The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
04/12/1995

Recalls and Field Corrections:  Foods -- Class I -- 04/12/1995

April 12, 1995                                                  95-15

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        Portland Light brand Frozen Rock Crab Salad Crabmeat, in 5
               pound boxes.  Recall #F-518-5.
CODE           None.  
MANUFACTURER   Portland Shellfish, Inc., Portland, Maine.
RECALLED BY    Portland Shellfish Sales, Inc., Lynn, Massachusetts, by
               telephone March 3, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Massachusetts.
QUANTITY       Firm estimates none remains on the market.
REASON         Product is contaminated with Listeria monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Osem brand Bamba Snacks, in 1.0 and 3.5 ounce packages. 
               Recall #F-527-5.
CODE           All codes.
MANUFACTURER   OSEM Food Industries Ltd., Petach-Tikva, Israel.
RECALLED BY    OSEM USA, Inc., Englewood Cliffs, New Jersey, by telephone
               February 23, 1995, followed by letter February 24, 1995. 
               Firm issued press release on February 24, 1995.  See also
               FDA statement dated February 10, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   California, Florida, Illinois, Louisiana, Massachusetts,
               Maryland, Michigan, Minnesota, North Carolina, New Jersey,
               New York, Pennsylvania, Texas.
QUANTITY       Approximately 1,175 cases were distributed.
REASON         Product is contaminated with Salmonella agona.

_______________
PRODUCT        Zetov brand Bamba Snakcs, in 1.0 and 3.5 ounce packages,
               product of Israel.  Recall #F-528-5.
CODE           All codes.
MANUFACTURER   OSEM Food Industries, Ltd., Petach-Tikva, Israel.
RECALLED BY    Zetov, Inc., Brooklyn, New York, by telephone February 22,
               1995, followed by letter and press release March 1, 1995. 
               See also FDA statement dated February 10, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   New York.
QUANTITY       2,954 cases (48 1-ounce packages) and 495 cases (24 3.5-
               ounce packages) were distributed; firm estimates none
               remains on the market.
REASON         Product is contaminated with Salmonella agona.

_______________
PRODUCT        Wild Fruit Candy, sugar free, salt free, 3.85 ounces.
               Recall #F-604-5.
CODE           4138.
MANUFACTURER   Sorbee International, Philadelphia, Pennsylvania.
RECALLED BY    King Nut Company, Solon, Ohio (repacker), by relabeling
               product September 15, 1994.  Firm-initiated field correction
               complete.
DISTRIBUTION   New York.
QUANTITY       10 cases (24 bags per case) were distributed.
REASON         Product contained undeclared FD&C Yellow No. 5.

_______________
PRODUCT        Rolets Real Cotton Candy, in 1 ounce plastic bags.  
               Recall #F-605-5.
CODE           All 1 ounce size bags shipped prior to 2/21/95 and coded
               5/01 and older. 
MANUFACTURER   Rolet Food Products Company, Brooklyn, New York.
RECALLED BY    Manufacturer, by letters of February 21, 1995, and March 17,
               1995.  Firm-initiated field correction (relabeling) ongoing.
DISTRIBUTION   Connecticut, Massachusetts, Maryland, Maine, New Hampshire,
               New Jersey, New York, Pennsylvania, Rhode Island, Virginia,
               Vermont.
QUANTITY       500,000 1-ounce packages were distributed; firm estimates
               none remains on the market.
REASON         Product contains undeclared FD&C Yellow No. 5, Red No. 3 and
               Red No. 40.


                                    -2-RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Quaker Sweet Crunch, crisps of corn cereal, packaged in 16
               ounce cellophane-type bags.  Recall #F-523-5.
CODE           Best before dates of 9/14/95-11/4/95
MANUFACTURER   Quaker Oats Company, Cedar Rapids, Iowa.
RECALLED BY    Quaker Oats Company, Chicago, Illinois, by fax and E-mail
               February 1, 1995, followed by visit.  Firm-initiated recall
               complete.
DISTRIBUTION   Tennessee, Illinois, Florida, Missouri, Texas, Pennsylvania,
               Iowa, California.
QUANTITY       Approximately 30,000 cases (12 bags per case) were
               distributed.
REASON         The product has an off odor.

_______________
PRODUCT        Van Camp's Red Beans, in 15-1/2 ounce cans.  
               Recall #F-602-5.
CODE           Can code:  MRB07/07I4A **** (where *** indicates the
               military time);
               Case Code: SEP 07 94 MA and SEP 07 94 MB.
MANUFACTURER   Morgan Foods, Inc., Austin, Indiana.
RECALLED BY    The Quaker Oats Company, Chicago, Illinois, by telephone
               March 9, 1995, followed by fax March 10, 1995.  Firm-
               initiated field correction (relabeling) ongoing.
DISTRIBUTION   Missouri, Nebraska.
QUANTITY       181 cases (24 cans per case) of mislabeled product were
               distributed; firm estimated that 25 cases remained on market
               at time of recall initiation.
REASON         The product is mislabeled in that cans of chicken pasta soup
               were labeled as containing red beans.

_______________
PRODUCT        Powdered Non-Dairy Creamers, packaged in 6 ounce, 11 ounce,
               16 ounce, 20 ounce, and 22 ounce retail containers.  Also
               packaged in 50 pound, 1,000 pound and 1,100 pound commercial
               totes, distributed under the following labels:  Cremora,
               Cremora Lite, America's Choice, Astor, A & P, Bi-Lo, Laura
               Lynn, Laura Lynn Lite, Piggly Wiggly, Rainbow, Shurfine,
               Starflake, President's Choice, and Sun Spin.  
               Recall #F-603-5.
CODE           z21may96, z23may96, z28may96, z20nov95 (on retail
               containers); C4140, C4141, C4143-C4148 (on commercial
               totes).
MANUFACTURER   Borden, Inc., Chester, South Carolina.
RECALLED BY    Borden, Inc., Columbus, Ohio, by letters dated September 23,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Canada, Costa Rica, Singapore.
QUANTITY       Approximately 65,000 retail cases, and approximately 1,000
               commercial totes were distributed.
REASON         Product contains small thread-like pieces of metal.


                                    -3-RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        A/B (Antipyrine/Benzocaine) Otic Drops, Rx, in 10 ml and 1/2
               fluid ounce glass bottles, under the following labels: 
               Goldline, H.L. More, Aurodex, G PR, Rugby, Akorn, Clay Park. 
               Recall #D-106-5.
CODE           For the 10ml: 
               Lot #     EXP       Lot #     EXP    
               S412      4/95      U942      2/96       
               S678      5/95      V089      3/96                 
               T178      7/95      V244      3/96                 
               T199      7/95      V259      4/96                 
               T450      8/95      V569      5/96                 
               V575      5/96      X003      11/96                
               X012      11/96      
               For the 1/2 fluid ounce: 
               Lot #     EXP       Lot #     EXP       Lot #     EXP    
               P007      4/95      S405      4/96      U704      11/97      
               P013      4/95      S412      4/96      U942      2/97 
               P274      5/95      S185      3/96      V089      3/97       
               P606      7/95      S688      5/96      V575      5/97       
               P806      8/95      S909      6/96      V259      4/97       
               R724      1/96      S917      6/96                           
               R830      1/96      T178      7/96                           
               S178      3/96      T966      10/96      
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY5   Manufacturer, by letter April 3, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Virgin Islands, West Indies, Guyana.
QUANTITY       Approximately 534,183 10-ml bottles and 800,797 15-ml glass
               bottles were distributed.
REASON         Product does not meet pH and water content specifications.

_______________
PRODUCT        Potassium Phosphates Injection, USP, in 5 ml and 50 ml
               single dose vials, used as a source of phosphorus Recall #D-
               107-5.
CODE           Lot numbers:  4399 (5 ml) EXP 5/96, 4531 (50 ml) EXP 7/96.
MANUFACTURER   Luitpold Pharmaceuticals, Inc., Shirley, New York.
RECALLED BY    Manufacturer, by letter March 16, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       115,325 vials of lot 4399 and 26,450 vials of lot 4531.
REASON         Presence of crystallization.

_______________
PRODUCT        Rectacaine, and all other private labels of the firm's
               Hemorrhoidal Suppositories containing "Live Yeast Cell
               Derivatives" (LYCD).  Recall #D-108-5.
CODE           Lot #2120 EXP 1/96.
MANUFACTURER   Dr. Rose, Inc., Madison, Connecticut.

                                    -4-RECALLED BY    The Reese Chemical Company, Cleveland, Ohio, by letter March
               28, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       323 packages containing 12 suppositories each were
               distributed.
REASON         Product distributed after September 1994 ruling regarding
               live yeast cell derivative (LYCD) content.

_______________
PRODUCT        Bisacodyl Suppositories, 10 Mg, USP.  Non-prescription
               laxative, in aluminum foil packages of 8s, 12s, 50s, 100s,
               500s and 1000s, marketed under firm's own label and the     
               following private labels: 
               -SUPERX - Bisacodyl Suppositories U.S.P., Distributed by     
               SUPERX DRUGS Cincinnati, Ohio 45246 
               -REVCO Bisacodyl U.S.P. LAXATIVE (bisacodyl suppositories)   
               Dist. by REVCO D.S., Inc. Twinsburg, Ohio 44087 
               -RUGBY Bisacodyl Suppositories, Dist. by Rugby Laboratories, 
               Inc., Rockville Centre, N.Y. 11570 
               -BIG B DISCOUNT DRUGS Gentle Laxative Suppositories
               (Bisacodyl USP) Distributed By: BIG B DISCOUNT DRUGS,
               Birmingham, AL.  Recall #D-109-5. 
CODE           Batch#    Exp.    Batch#    Exp.    Batch#    Exp. 
               M250      4/95     N091    7/95     O084     11/95       
               M303      4/95     N140    7/95     O085     11/95 
               M304      4/95     N245    8/95     0086     11/95 
               M428      4/95     N246    8/95     O087     11/95 
               M429      4/95     N247    8/95     O093     11/95 
               M430      4/95     N308    8/95     O254     12/95 
               Batch#    Exp.    Batch#    Exp.    Batch#    Exp.      
               M584      5/95     N309    8/95     O255     12/95 
               M585      5/95     N310    8/95     O256     12/95 
               M586      5/95     N366    8/95     O257     12/95 
               M588      5/95     N367    8/95     O258     12/95 
               M589      5/95     N368    8/95     O259     12/95 
               M687      5/95     N447    8/95     O304      1/96 
               M688      5/95     N448    8/95     O305      1/96 
               M689      5/95     N449    8/95     O306      1/96 
               M793      6/95     N521    9/95     O363      2/96 
               M794      6/95     N522    9/95     O364      2/96 
               M795      6/95     N523    9/95     O614      2/96 
               M856      6/95     N700    10/95    O619      3/96 
               M857      6/95     N701    10/95    O823      3/96 
               M858      6/95     N702    10/95    O824      3/96 
               M953      6/95     N703    10/95    O825      3/96 
               M954      6/95     N722    10/95    O826      3/96 
               M955      6/95     N723    10/95    O827      3/96 
               M956      6/95     N724    10/95    O931      3/96 
               M960      6/95     N725    10/95    O974      3/96 
               M961      7/95     N950    11/95    O975      3/96 
               N089      7/95     N951    11/95    O977      3/96 
               N090      7/95     N952    11/95    O978      3/96 

                                    -5-               P315      5/96     R879    1/97     T476      8/97 
               P394      5/96     R880    1/97     T477      8/97 
               P588      6/96     R881    1/97     T478      9/97 
               P589      6/96     R882    1/97     T628      9/97 
               P590      6/96     R933    1/97     T629      9/97 
               P591      6/96     R934    1/97     T630      9/97 
               P592      7/96     R935    2/97     T631      10/97 
               P675      7/96     S333    4/97     T632      10/97 
               P676      7/96     S334    4/97     T950      10/97 
               P734      7/96     S335    4/97     T951      10/97 
               P735      7/96     S336    4/97     T952      10/97 
               P736      7/96     S337    4/97     T953      10/97 
               P867      8/96     S647    5/97     U370      11/97 
               P868      8/96     S648    5/97     U371      11/97 
               P869      8/96     S649    5/97     U375      11/97 
               R007      9/96     S650    5/97     U376      11/97 
               R008      9/96     S651    5/97     U585      12/97 
               R009      9/96     S764    5/97     U586      1/98 
               R010      9/96     S765    6/97     U587      1/98 
               R050      9/96     S766    6/97     U701      1/98 
               R052      9/96     S854    6/97     U702      1/98 
               R110      9/96     S855    6/97     U843      1/98 
               R111      9/96     S856    6/97     U844      2/98 
               R112      9/96     S951    6/97     U845      3/98 
               R350      10/96    S998    6/97     V080      3/98 
               R351      10/96    T043    6/97     V081      3/98 
               R352      10/96    T203    7/97     V401      4/98 
               R353      10/96    T204    7/97     V402      4/98 
               R521      11/96    T205    7/97     V506      5/98 
               R774      12/96    T270    7/97     V507      5/98 
               R775      12/96    T271    7/97     V508      5/98 
               R776      12/96    T272    7/97     V592      5/98 
               R777      1/97     T273    8/97     V593      5/98 
               R878      1/97     T475    8/97     V676      5/98 
               V677      5/98     V681    6/98     V800      6/98 
               V678      5/98     V682    6/98     V801      6/98 
               V679      5/98     V683    6/98     W054      7/98 
               V680      5/98     V684    6/98     W055      7/98 
               M689      5/95     O613    2/96     O364      2/96 
               P736      7/96 
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter to be mailed on April 17, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       A total of 758,339 labeled packages were distributed to date 
               (427,466/8s; 272,014/12s; 7,526/50s; 49,986/100s;
               1,171/500s; 176/1000s).  This product was also sold in bulk
               to two customers (E-Z EM Inc., and P. F. Laboratories).  
REASON         Subpotency.

                                    -6-_______________
PRODUCT        Supac Analgesic Tablets, in 100 and 1000 count bottles, OTC
               pain reliever.  Recall #D-110-5.
CODE           Lot 4J21 EXP 9/97.
MANUFACTURER   Mission Pharmacal Company, San Antonio, Texas.
RECALLED BY    Manufacturer, by letter March 23, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Sri Lanka.
QUANTITY       2,964 100-tablet and 299 1000-tablet bottles were
               distributed; firm estimated that 20-30 percent of product
               remained on market at time of recall.
REASON         Product does not meet uniformity specifications for the
               caffeine ingredient.

_______________
PRODUCT        Hemorrhoidal A Suppositories, 2.08 grams, packed in 12s,
               24s, 100s, and 1000s aluminum foil blister packages under
               the Clay Park Labs and Eckerd Drug labels, a
               non-prescription drug.  Recall #D-111-5.
CODE            Lot #       EXP           Lot #        EXP 
                S704        06/95         U746         01/96 
                T004        06/95         U847         01/96 
                T133        07/95         U984         02/96 
                T775        09/95         V319         04/96 
                T949        10/95         V591         05/96 
                U232        10/95         V996         07/96 
                U409        11/95         W267         08/96 
                U410        11/95         W515         09/96 
                U584        12/95         W993         11/96 
                U699        01/96         X133         12/96  
                U700        01/96.  
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York. 
RECALLED BY    Manufacturer, by letter March 30, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   New York, New Jersey, California, Ohio, Florida, Georgia,
               Texas, Pennsylvania, Mississippi, Tennessee, Massachusetts,
               Arkansas, Alabama, Indiana, Ohio, Puerto Rico, Hong Kong. 
QUANTITY       Approximately 75,185 units were distributed.
REASON         Stability data does not support expiration date.


HUMAN TISSUE MANDATORY
RECALLS:============================================
_______________
PRODUCT        Human Tissue From Cadaveric Donors, indicated for human
               transplant.  Recall #B-198-5.
CODE           Donor numbers:  
               AK9408002, BA9303001, BA9303002, BA9304001, BA9304003,
               BA9305004, BA9305006, BA9306001, BA9306003, BA9306004,
               BA9306005, BA9306006, BA9307008, BA9307010, BA9307011,
               BA9308007, BA9308008, BA9308009, BA9308010, BA9309002,
               BA9309003, BA9309004, BA9309005, BA9309006, BA9309008,
               BA9309009, BA9310009, BA9310012, BA9310013, BA9310014,

                                    -7-               BA9310015, BA9310016, BA9310017, BA9310018, BA9311001,
               BA9311002, BA9311003, BA9311005, BA9311006, BA9311007,
               BA9311008, BA9311009, BA9312001, BA9312003, BA9312004,
               BA9312005, BA9312006, BA9312007, BA9312008, BA9312009,
               BA9312010, BA9312011, BA9312013, BA9401002, BA9401005,
               BA9401006, BA9401008, BA9401009, BA9402003, BA9402004,
               BA9402005, BA9402007, BA9402008, BA9402009, BA9402010,
               BA9402011, BA9402012, BA9402013, BA9403002, BA9404001,
               BA9406007, BA9406008, BA9406008, BA9406009, BA9406009,
               BA9406010, BA9406011, BA9406012, BA9406013, BA9408004,
               BA9408005, BA9408006, BA9408008, BA9408010, BA9408012,
               BA9408013, BA9408014, BA9408015, BA9409001, BA9409002,
               BA9409003, BA9409004, BA9409006, BA9409007, CA9104098,
               CA9108099, CA9201043, CA9203102, CA9204093, CA9208081,
               CA9209013, CA9209021, CA9209072, CA9304043, CA9304085,
               CA9305097, CA9306011, CA9306074, CA9308061, CA9309110,
               CA9309115, CA9310150, CA9312012, CA9312016, CA9401054,
               CA9401082, CA9401110, CA9402006, CA9402091, CA9403063,
               CA9403075, CA9403109, CA9406110, CA9407057, CA9408047,
               CA9409017, CA9409108.
MANUFACTURER   Lions Doheny Eye and Tissue Bank, Los Angeles, California.
RECALLED BY    Manufacturer, by letters dated November 1, 1994, and
               December 5, 1994.  FDA ordered recall ongoing.
DISTRIBUTION   California, Florida, Indiana, New Jersey, New York, Georgia,
               Washington, D.C., Argentina, Canada, Egypt, France, Greece,
               Israel, Italy. 
QUANTITY       All lots involved.
REASON         Human tissue from cadaveric donors was distributed which: 
               (1) lacked adequate documentation about the donors' relevant
               medical history related to risk factors for, or clinical
               evidence of, hepatitis B, hepatitis C, or human
               immunodeficiency virus (HIV) infection; or (2) the donor
               tested repeatedly reactive for hepatitis B surface antigen
               (HBsAg).


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-195/196-5.
CODE           Unit #2982243.
MANUFACTURER   Central Florida Blood Bank, Inc., Orlando, Florida.
RECALLED BY    Manufacturer, by letter July 1, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Florida.
QUANTITY       1 unit of each component.
REASON         Blood products, which were collected from a donor who had a
               history of nose piercing within 12 months of donation, were
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  
               Recall #B-201/202-5.

                                    -8-CODE           Unit #2130944.
MANUFACTURER   Manatee Community Blood Center, Inc., Brandenton, Florida.
RECALLED BY    Manufacturer, by letter dated November 29, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Florida.
QUANTITY       1 unit of each component.
REASON         Blood products, which tested initially reactive for the
               hepatitis B surface antigen (HBsAg) followed by a single
               repeat negative test for HBsAg, were distributed.

_______________
PRODUCT        Fresh Frozen Plasma.  Recall #B-203-5.
CODE           Unit #L24042.
MANUFACTURER   South Bend Medical Foundation, Inc., South Bend, Indiana.
RECALLED BY    Manufacturer, by telephone August 4, 1992.  Firm-initiated
               recall complete.
DISTRIBUTION   Indiana.
QUANTITY       1 unit.
REASON         Blood product, which tested repeatedly reactive for the
               antibody to the human immunodeficiency virus type 1 and 2
               (anti-HIV-1/2), was distributed.

_______________
PRODUCT        Whole Blood CPDA-1.  Recall #B-204-5.
CODE           Unit 5012683.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City,
               Missouri.
RECALLED BY    Manufacturer, by letter dated May 25, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Missouri.
QUANTITY       1 unit.
REASON         Blood product, which tested negative for the antibody to the
               human immunodeficiency virus type 1(anti-HIV-1) but was
               collected from a donor who previously tested repeatedly
               reactive to anti-HIV-1, was distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-206-5.
CODE           Unit #P01327.
MANUFACTURER   South Bend Medical Foundation, Inc., South Bend, Indiana.
RECALLED BY    Manufacturer, by telephone June 29, 1992.  Firm-initiated
               recall complete.
DISTRIBUTION   Indiana.
QUANTITY       1 unit.
REASON         Blood product, which tested negative for the antibody to the
               human immunodeficiency virus type 1 and 2 (anti-HIV-1/2) but
               was collected from a donor who previously tested repeatedly
               reactive for anti-HIV-1/2, was distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-207-5.
CODE           Unit #R20646.

                                    -9-MANUFACTURER   Walter L. Shepeard Community Blood Center, Augusta, Georgia.
RECALLED BY    Manufacturer, by letter November 7, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit.
REASON         Blood product, which tested negative for the antibody to the
               hepatitis B core antigen (anti-HBc) but was collected from a
               donor who previously tested repeatedly reactive for anti-
               HBc, was distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-208/209-5.
CODE           Unit #E06025.
MANUFACTURER   Walter L. Shepeard Community Blood Center, Augusta, Georgia.
RECALLED BY    Manufacturer, by letter February 10, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit of each component.
REASON         Blood products, which tested initially reactive for the
               antibody to the hepatitis B core antigen (anti-HBc), were
               distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-218-5.
CODE           Unit #R96189.
MANUFACTURER   Mississippi Valley Regional Blood Center, Davenport, Iowa.
RECALLED BY    Manufacturer, by telephone August 25, followed by letter
               November 23, 1994.  Firm-initiated recall complete
DISTRIBUTION   Illinois.
QUANTITY       1 unit.
REASON         Blood product, which was improperly tested for the antibody
               to the human immunodeficiency virus type 1 (anti-HIV-1), was
               distributed.


RECALLS AND FIELD CORRECTIONS -- BIOLOGICS -- CLASS III
===================
_______________
PRODUCT        Source Plasma.  Recall #B-205-5.
CODE           Unit NRD529.
MANUFACTURER   Sera-Tec Biologicals, Inc., Fort Wayne, Indiana.
RECALLED BY    Manufacturer, by telephone March 31, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   North Carolina.
QUANTITY       1 unit.
REASON         Source Plasma, which was collected from a donor who
               previously tested positive for syphilis, was distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-210/211-5.
CODE           Unit #1141915.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City,
               Missouri.

                                   -10-RECALLED BY    Manufacturer, by letter May 25, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Missouri, Kansas.
QUANTITY       1 unit of each component.
REASON         Blood products, which were collected from a donor who tested
               initially negative for the antibody to the hepatitis B core
               antigen (anti-HBc) but on a subsequent donation tested
               repeatedly reactive for anti-HBc, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Red Blood Cells - Leukocytes
               Removed; (c) Platelets; (d) Fresh Frozen Plasma; (e)
               Recovered Plasma.  Recall #B-212/216-5.
CODE           Unit numbers:  (a) 4449525, 4449527, 2259316, 4143774,
               4143785, 4450305, 4450369, 4450457; (b) 1903110, 4450307;
               (c) 4449525, 4449527, 1903110, 4143774, 4450303, 4450306,
               4450310, 4450363, 4450364, 4450369, 4450403, 4450412; (d)
               1903110; (e) 4449525, 4449527, 4143785, 4143774.
MANUFACTURER   Florida Georgia Blood Alliance, Inc., Jacksonville, Florida.
RECALLED BY    Manufacturer, by telephone, followed by letter dated
               December 12, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Ohio, Florida, Massachusetts, California, Tennessee.
QUANTITY       (a) 8 units; (b) 2 units; (c) 12 units; (d) 1 unit; (e) 4
               units.
REASON         Blood products, which were collected from donors whose
               health history screening were inadequately performed, were
               distributed.

_______________
PRODUCT        Platelets.  Recall #B-217-5.
CODE           Unit #1186273.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City,
               Missouri.
RECALLED BY    Manufacturer, by telephone August 21, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Missouri.
QUANTITY       1 unit.
REASON         Platelets which were not stored under continuous agitation
               were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Siemens MEVASIM Radiation Therapy Simulator Systems with
               plastic rotatable cassette holders, teletherapy planning
               systems for the treatment of cancer:
               (a) MEVASIM Radiation Therapy Simulator System;
               (b) MEVASIM 2 Radiation Therapy Simulator System;
               (c) MEVASIM S Radiation Therapy Simulator System;
               (d) MEVASIM S/27 Radiation Therapy Simulator System.
               Recall #Z-335/338-5.
CODE           All serial numbers.

                                   -11-MANUFACTURER   Mecaserto, St. Thibault-des-Vigenes, France.
RECALLED BY    Siemens Medical Systems, Inc., Concord, California, by
               letters mailed week of May 23, 1994.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       259 units are subject to field correction.
REASON         Assemblies used to attach the rotatable cassette holder to a
               MEVASIM system may fail causing the cassette holder assembly
               to fall.

_______________
PRODUCT        5 in 1 Sterile Plastic Tubing Connectors, single use.
               Recall #Z-601-5.
CODE           Lot numbers:  9776, 9777, 9778, 9779 representing all lots.
MANUFACTURER   MCP Industries, Inc., Decatur, Alabama.
RECALLED BY    Sterile Design, Division of Maxxim Medical, Clearwater,
               Florida, by letter February 14, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Alabama, Arkansas, California, Florida, Georgia,
               Massachusetts, Mississippi, Missouri, New Jersey, North
               Carolina, Pennsylvania, South Carolina, Tennessee, Texas,
               West Virginia.
QUANTITY       Approximately 216,000 devices were manufactured and are
               subject to recall.
REASON         The sterility of the connectors has been compromised due to
               inadequate documentation to assure adequate sterilization.

_______________
PRODUCT        MP Video, Medicam 900 Insufflator Sterile Disposable Filter
               and Tubing Set.  Recall #Z-602-5.
CODE           Product #01-5212, Lot #AF4L0002.
MANUFACTURER   Cybermedic, Boulder, Colorado.
RECALLED BY    MP Video, Medway, Massachusetts, by telephone or by fax
               January 26, 1995, followed by letter January 27, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Connecticut, Florida, Illinois, Massachusetts, Maine, North
               Carolina, New Hampshire, New York, Texas, Utah, Wisconsin,
               Chile, Spain, Colombia, Belgium.
QUANTITY       63 cases (10 bags per case) were distributed.
REASON         The sterility of the device has been compromised as
               evidenced by loss of package integrity.

_______________
PRODUCT        Sterile Design Custom Surgical Procedure Trays containing
               Coated VICRYL Sutures  (NOTE:  Only the referenced sutures
               made by Ethicon are under recall NOT the entire surgical
               tray).  Recall #Z-603/635-5.
                                            MOD/CAT# 
CODE           Akron City Hosp              87/ACH01AO, 88/ACH01AO,         
                                            89/ACH01AO 

                                   -12-               Baptist Hospital Nashville   70/BH33NT, 68/BH33NT, 71/BH33NT
               Baptist Med Ctr              48/BM04KCM, 46/BM04KCM 
                                            47/BM04KCM, 33/BMC12OCO 
                                            31/BMC12OCO, 32/BMC12OCO 
                                            35/BMC12OCO, 18/BMC23OCO 
                                            19/BMC23OCO, 34/BMC12OCO 
               Community Hosp               03/CH07MI, 03/CH09MI 
                                            03/CH08MI, 04/CH08MI 
                                            03/CH09MI, 04/CH07MI 
                                            04/CH09MI, 05/CH08MI 
               Cuyahoga Falls Hosp          03/CFGH17CFO, 04/CFGH17CFO 
                                            05/CFGH15CFO, 07/CFGH10CFO 
                                            08/CFGH10CFO, 03/CFGH18CFO 
                                            06/CFGH12CFO, 04/CFGH15CFO 
                                            02/CFGH18CFO 
               Cuyahoga Falls Hosp          04/CFGH18CFO, 01/CFGH19CFO 
               Dr. Michael Katz             17/MK01VF 
               Elyria Memorial              02/EMH03EC 
               Geisinger Med Ctr            30/GMC09DP, 57/GMC04DP 
                                            21/GMC10DP, 22/GMC10DP 
                                            40/GMC06DP 
               Gillette Childrens           18/GCH06SPM 
               Heb Harris Med               94/HEB01BT 
               Heb Harris Meth              19/HEB18BT, 20/HEB18BT 
                                            92/HEB01BT, 93/HEB01BT 
               Hillcrest Hospital           14/HH12MHO, 15/HH12MHO 
               Iowa Methodist               02/IMM02DI, 03/IMM02DI 
               Jennie Edmundson             26/JEH02CBI 
               Jewish Hospital              12/JH01LK, 13/JH01LK, 14/JH01LK
               Lutheran General             01/LGH25PRI, 02/LGH25PRI 
                                            03/LGH25PRI 
               Mendocino Coast              32/MCH01FBC, 33/MCH01FBC 
               Menorah Medical Ctr          69/MMC10KCM, 50/MMC12KCM 
                                            51/MMC12KCM, 70/MMC10KCM 
                                            49/MMC12KCM, 68/MMC10KCM 
               Methodist Southlake          52/MH03MI, 53/MH03MI, 86/MH10MI
                                            37/MH13MI 
               Mississippi Meth             25/MMRC05JM, 24/MMRC07JM 
                                            26/MMRC05JM 
               Missouri Baptist             94/MBHO4SLM, 95/MBH04SLM 
                                            96/MBH04SLM 
               N.C. Baptist Hosp            38/NCBH03WSNC, 86/NCBH01WSNC 
                                            87/NCBH01WSNC, 39/NCBH03WSNC 
                                            88/NCBH01WSNC, 40/NCBH01WSNC 
                                            90/NCBH01WSNC 
               Nix Med Ctr                  14/NMC06SAT, 15/NMC06SAT 
               Sentara Leigh Hosp           71/LMH04NV 
               St. Anthony Med Ctr          49/SAMC11SLM 
               St. Catherine's Hosp         62/SCH13ECI 
               St. Clare Hosp               01/SCH09TW 
               St. Joseph Hosp MI           04/SJM04MI 
               St. Joseph Hosp              01/SJH11TF 

                                   -13-               St. Mary Med Ctr             57/SMMC05HI, 58/SMMC05HI 
                                            11/SMMC06HI, 56/SMMC05HI 
               St. Marys Hosp               18/SMH20RV, 19/SMH20RV 
                                            22/SMH18RV, 20/SMH20RV 
                                            07/SMH19RV 
               Trinity Luth Hosp            78/TLH01KCM, 79/TLH01KCM 
               Williamsport Hosp            13/WHP03WP.
MANUFACTURER   Sterile Design, Clearwater, Florida.
RECALLED BY    Manufacturer, letters of September 16 and 19, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,873 trays were distributed.
REASON         The sterility of the sutures contained in these devices
               cannot be assured because of possible microorganism
               contamination during aseptic processing, following
               sterilization.

_______________
PRODUCT        IV Extension Set with Add-on Filter, indicated for the IV
               infusion of medical fluids.  Recall #Z-636-5.
CODE           Catalog #V5812, lot #F4K307.
MANUFACTURER   McGaw, Inc., Sabana Grande, Puerto Rico.
RECALLED BY    McGaw, Inc., Irvine, California, by letter December 22,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       643 cases were distributed.
REASON         Due to an inadequate solvent bond, the product leaks between
               the tubing and filter assembly, which could lead to under-
               infusion and blood loss.

_______________
PRODUCT        Groshong Catheter Trays:  (a) 9.5 French Dual Lumen Groshong
               Catheter Tray; (b) 7 French Single Lumen Groshong Catheter
               Tray.  Recall #Z-640/641-5.
CODE           Product No.       Lot No.
               (a) 7726952      36IE1909
               (b) 60366-4      36IE1909.
MANUFACTURER   Bard Access Systems, Inc., Salt Lake City, Utah.
RECALLED BY    Manufacturer, by telephone November 22, 1994, followed by
               letter November 28, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Alabama, Arizona, California, Colorado, Florida, Illinois,
               Michigan, New Hampshire, New Jersey, Oklahoma, Texas, West
               Virginia.
QUANTITY       64 units were distributed.
REASON         Mislabeling -- Single lumen catheters were inadvertently
               mixed in with dual lumen catheters.

_______________
UPDATE         Recall #Z-440/441-5, Zinetics 24 Single and Dual Channel
               Internal Reference Catheters, manufactured by Zinetics
               Medical, Inc., Salt Lake City, Utah, which appeared in the
               March 8, 1995 Enforcement Report has been extended to 

                                   -14-               include  additional lot numbers:
               Z-440-5 - Part #15-245, Lot #Z910;
               Z-441-5 - Part #15-247-15, Lot numbers:  Z917, Z963.

_______________
UPDATE         Recall #Z-471-5, Pegasys 2.1 Software System, manufactured
               by ADAC Laboratories, Milpitas, California, which appeared
               in the March 15, 1995 Enforcement Report should read:
               REASON:  An error in the software could cause the upper
               portion of the image to appear brighter than it should be. 
               The problem could lead to a misdiagnosis.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Rota-Stat, in-vitro diagnostic test kits, used to detect
               rotavirus's.  Recall #Z-638-5.
CODE           Lot #12094.
MANUFACTURER   Eldan Technologies Company, Ltd., Jerusalem, Israel.
RECALLED BY    Isolab, Inc., Norton, Ohio, by letter February 1, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       41 kits were distributed.
REASON         The control latex in this kit is clumping or agglutinating.

_______________
PRODUCT        Sportrol Growth Promotion Test Suspension (Candida
               albicans), used to test the growth properties of prepared
               bacteriological media.  Recall #Z-639-5.
CODE           Lot #GCA009-08.
MANUFACTURER   NAmSA, Northwood, Ohio.
RECALLED BY    Manufacturer, by telephone, followed by letter dated January
               30, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Japan.
QUANTITY       35 vials were distributed.
REASON         The growth promotion test suspension is contaminated with
               Bacillus species.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS I
============
_______________
PRODUCT        Complete Feed, custom batch.  Recall #V-025-5.
CODE           Bulk custom mix.
MANUFACTURER   Coldwater Grain Company, Coldwater, Ohio.
RECALLED BY    Premiere Agri Technologies, Inc., doing business as Henwood
               Feed Additives, Fort Wayne, Indiana, by telephone May 10,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Ohio.
QUANTITY       1 3-ton batch was distributed.  Product has been reworked
               and relabeled.
REASON         Superpotent for Arsanilic Acid.


                                   -15-RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
===========
_______________
PRODUCT        Kent 18% Golden Nugget ASP 250 Medicated Type C Category II
               pelleted feed, in 50-pound bags.  Recall #V-027-5.
CODE           Lot #R218A.
MANUFACTURER   Kent Feeds, Inc., Rockford, Illinois.
RECALLED BY    Manufacturer, by telephone February 10-13, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, Wisconsin.
QUANTITY       395 bags were distributed; firm estimates none remains on
               the market.
REASON         Feed was subpotent for penicillin.

                                   -16-


END OF ENFORCEMENT REPORT FOR APRIL 12, 1995.  BLANK PAGES MAY
FOLLOW.



                                   ####