The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

November 3, 1999                                               99-44

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS II

_______ PRODUCT Liquid Pasteurized Whole Eggs in white colored square shaped plastic buckets with a paper label affixed to the exterior of the bucket, 30, pounds and in plastic drums of approximately 55- gallon volume size with black plastic lids with each drum containing 450 pounds of product. Recall #F-004-0. CODE The lot number is 196-9 that reflects product manufactured on 7/15/99. MANUFACTURER Hi Point Industries, L.L.C., Norco, California. RECALLED BY Manufacturer, by telephone on July 21, 1999. Firm-initiated recall complete. DISTRIBUTION California. QUANTITY 14 450-pound drums or 6,300 pounds and 1,180 30-pound buckets or 35,400 pounds were distributed. REASON Product is contaminated with Salmonella C-1. ________ PRODUCT PowerAde Mountain Blast in 11.5 ounce cans. Recall #F-009-0. CODE Lot 9203MD. MANUFACTURER The Minute Maid Company, for Coca-Cola USA at Coca-Cola Bottling Company of Chicago, Niles, Illinois. RECALLED BY Coca-Cola USA Operations, Atlanta, Georgia, by E-mail or fax on August 18, 1999. Firm-initiated recall ongoing. DISTRIBUTION Alabama, Florida, Georgia, Kansas, Louisiana, Tennessee, Wisconsin. QUANTITY 5.346 cases (24 cans per case) were distributed. REASON Some cans actually contain Lemon-Lime flavor. The Lemon-Lime flavor contains FD&C Yellow No. 5 Lake and brominated vegetable oil that are not listed as ingredients on the Mountain Blast flavor.

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS III

________ PRODUCT AMROS, Pure Natural, Mixed Nuts Honey, in 14 ounce jars. Recall #F-008-0. CODE None. MANUFACTURER AMROS The Second, Inc., Somerset, New Jersey. RECALLED BY Manufacturer, by letter sent on August 25, 1999. Firm-initiated recall complete. DISTRIBUTION Minnesota, Missouri, Washington state. QUANTITY 6 cases (12 jars per case) were distributed. REASON Product does not bear an ingredient statement. ________ PRODUCT PowerAde Mountain Blast, in 20 ounce PET containers. Recall #F-012-0. CODE 9112CC. MANUFACTURER The Minute Maid Company for Coca-Cola USA at United Packers, Inc., Chattanooga, Tennessee. RECALLED BY Coca-Cola USA Operations, Atlanta, Georgia, by E-mail and fax on September 3, 1999. Firm-initiated recall ongoing. DISTRIBUTION Alabama, Arkansas, California, Florida, Georgia, Kentucky, Louisiana, Mississippi, New Mexico, North Carolina, Ohio, Oklahoma, South Carolina, Tennessee, Texas. QUANTITY 23,924 cases (24 units per case) were distributed; firm estimated that 2,471 cases remained on market at time of recall initiation. REASON Product is contaminated with mold.

RECALLS AND FIELD CORRECTINS: BIOLOGICS -- CLASS II

________ PRODUCT Blood Bank Control Systems. Recall #B-020-0. CODE Software versions 4.0 and greater. MANUFACTURER Blood Bank Computer Systems, Kent, Washington. RECALLED BY Manufacturer, by letter dated September 1, 1998. Firm-initiated field correction complete. DISTRIBUTION California, Florida, Illinois, Kansas, Pennsylvania, Texas, Washington State. QUANTITY 14 systems were distributed. REASON Computer software contains programming errors that could potentially result in the release of a unit that has not been tested for antibodies. ________ PRODUCT a) Red Blood Cells; b) Platelets; c) Recovered Plasma. Recall #B-021/023-0. CODE Unit #04R37916. MANUFACTURER American Red Cross, Dedham, Massachusetts. RECALLED BY Manufacturer, by letter dated June 3, 1999. Firm-initiated recall complete. DISTRIBUTION New York and Maine. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor with a history of cancer. ________ PRODUCT a) Red Blood Cells; b) Source Leukocytes for Manufacture. Recall #B-026/027-0. CODE Unit #0700265. MANUFACTURER Bergen Community Regional Blood Center, Paramus, New Jersey. RECALLED BY Manufacturer, by fax and/or letter on February 19, 1999. Firm- initiated recall complete. DISTRIBUTION New Jersey and New York. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor who reported travel to an area designed as endemic for malaria. ________ PRODUCT Red Blood Cells. Recall #B-032-0. CODE Unit #49W24412. MANUFACTURER American Red Cross Blood Services, Tulsa, Oklahoma. RECALLED BY Manufacturer, by letter dated July 13, 1999. Firm-initiated recall complete. DISTRIBUTION California. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who had lived in an area designated as endemic for malaria. ________ PRODUCT Red Blood Cells. Recall #B-033-0. CODE Unit #42S66951. MANUFACTURER American Red Cross Blood Services, Cleveland, Ohio. RECALLED BY Manufacturer, by telephone eon June 16, 1999. Firm-initiated recall complete. DISTRIBUTION Ohio. QUANTITY 1 unit was distributed. REASON Blood product was labeled as leukoreduced but did not undergo leukoreduction. ________ PRODUCT Red Blood Cells. Recall #B-047-0. CODE Unit #3008296. MANUFACTURER Wellmont Health System, doing business as Bristol Regional Medical Center, Bristol, Tennessee. RECALLED BY Manufacturer, by letter dated September 14, 1999. Firm-initiated recall ongoing. DISTRIBUTION North Carolina. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who reported travel to an area designated as endemic for malaria. ________ PRODUCT Red Blood Cells. Recall #B-048-0. CODE Unit #29F28068. MANUFACTURER American Red Cross Blood Services, Norfolk, Virginia. RECALLED BY Manufacturer, by letter dated February 16, 1999. Firm-initiated recall complete. DISTRIBUTION Virginia. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who reported having lived in an area designated as endemic for malaria. ________ PRODUCT Platelets, Pheresis, Leukoreduced, Irradiated. Recall #B-049-0. CODE Unit #16KF20408, part 2. MANUFACTURER American Red Cross Blood Services, Columbus, Ohio. RECALLED BY Manufacturer, by letter dated August 6, 1999. Firm-initiated recall complete. DISTRIBUTION Ohio. QUANTITY 1 unit was distributed. REASON Blood product was incorrectly labeled as HLA matched. ________ PRODUCT a) Whole Blood; b) Red Blood Cells; c) Red Blood Cells, Deglycerolized; d) Platelets; e) Platelets, Pheresis; f) Fresh Frozen Plasma; g) Fresh Frozen Plasma, Pheresis; h) Cryoprecipitated AHF; i) Recovered Plasma. Recall #B-050/058-0. CODE FDA, Center for Biologics Evaluation and Research, Contact Office of Compliance (301) 827-6220 for individual unit numbers recalled. MANUFACTURER New York Blood Services (Center West), New York, New York. RECALLED BY Manufacturer, by letter dated June 24, 1999, followed by telephone or fax on July 21, 1999. Firm-initiated recall complete. DISTRIBUTION New York, New Jersey, California, Switzerland. QUANTITY a) 3 units; b) 36 units; c) 1 unit; d) 12 units; e) 2 units; f) 12 units; g) 4 units; h) 2 units; i) 23 units were distributed. REASON Blood products may not have been tested in accordance with the manufacturer's instructions for either HCV or HbsAg. ________ PRODUCT a) Red Blood Cells, Irradiated; b) Platelets, Pheresis, Irradiated, Leukocytes Reduced. Recall #B-059/060-0. CODE Unit Numbers: a) X48625 and X51412; b) X46101. MANUFACTURER Aurora Area Blood Bank, Aurora, Illinois. RECALLED BY Manufacturer, by letter dated September 2, 1999. Firm-initiated recall complete. DISTRIBUTION Illinois. QUANTITY a) 2 units; b) 1 unit was distributed. REASON Blood products lacked documentation of irradiation. ________ PRODUCT Cryoprecipitated AHF. Recall #B-061-0. CODE Unit Numbers: 53KQ54716, 53KR41958, 53KR41963, 53V88521, 53KX35154, 53KX35155, 53KX35159, 53KX35160, 53KS38864, 53KS38867, 53Q73963. MANUFACTURER American Red Cross Blood Services, Baltimore, Maryland. RECALLED BY Manufacturer, by telephone on March 30, 1999, and by letter dated April 5, 1999. Firm-initiated recall complete. DISTRIBUTION Maryland and District of Columbia. QUANTITY 11 units were distributed. REASON Blood products were incorrectly processed. ________ PRODUCT Red Blood Cells. Recall #B-062-0. CODE Unit Numbers: 29K17294, 53GM21762, 04W07543. MANUFACTURER American Red Cross Blood Services, Baltimore, Maryland. RECALLED BY Manufacturer, by telephone on June 21, 1998. Firm-initiated recall complete. DISTRIBUTION Maryland. QUANTITY 3 units were distributed. REASON Blood products had been stored at unsuitable temperatures. ________ PRODUCT Red Blood Cells. Recall #B-064-0. CODE 53FL72908, 53FL72921, 53GL98643, 53GL98646, 53M80764, 53J49805, 53L07668, 53GL98638. MANUFACTURER American Red Cross Blood Services, Baltimore, Maryland. RECALLED BY Manufacturer, by telephone on December 1, 1998. Firm-initiated recall complete. DISTRIBUTION Maryland. QUANTITY 8 units were distributed. REASON Blood products were shipped under unsuitable temperatures. ________ PRODUCT Source Plasma. Recall #B-072-0. CODE Units Numbers: 0460093369, 0460093796, 0460094429, 0460094577, 0460095257, 0460095360, 0460096091, 0460096327, 0460096919, 0460097084, 0460098070, 0460098680, 0460099083, 0460099657, 0460099910, 0460100486, 0460101017, 0460101449, 0460101679, 0460102098, 0460102503, 0460102803, 0460103346, 0460103641, 0460104012, 0460104305, 0460105080, 0460105538, 0460111501, 0460112096, 0460112276, 0460112893, 0460113524, 0460113786, 0460114539, 0460114876, 0460115382, 0460115748, 0460116210, 0460116502, 0460117042, 0460117285, 0460118020, 0460118258, 0460118723, 0460119051, 0460119869, 0460120687, 0460121017, 0460121755, 0460122173, 0460122559, 0460123065, 0460123442, 0460124969, 0460125535, 0460125991, 0460126608, 0460127188, 0460127377, 0460128166, 0460128405, 0460129463, 0460129998, 0460130361, 0460131088, 0460131410, 0460133409, 0460135648, 0460136077, 0460136863, 0460137543, 0460138018, 0460138239, 0460139194, 0460140109, 0460140475, 0460141015, 0460141977, 0460142478, 0460143124, 0460144640, 0460145657, 0460146057, 0460147105, 0460147581, 0460148111, 0460148816, 0460149341, 0460149754, 0460152191, 0460153434, 0460153954, 0460155404, 0460156089, 0460089621, 0460090231, 0460090643, 0460091315, 0460091897, 0460092103, 0460092720, 0460092897, 0460106426, 0460106793, 0460107237, 0460107596, 0460108054, 0460108267, 0460109319, 0460109531, 0460110153, 0460110571, 0460110988, 0460069539, 0460069764, 0460071017, 0460071435, 0460071905, 0460072558, 0460072881, 0460074373, 0460081025, 0460081246, 0460081810, 0460081946, 0460082592, 0460082710, 0460083411, 0460083519, 0460084499, 0460084706, 0460085245, 0460085661, 0460086577, 0460086727, 0460087820, 0460088008, 0460088567, 0460088809, 0460089267, 0460132235, 0460132597. MANUFACTURER Nabi Biomedical Center, Dallas, Texas. RECALLED BY Manufacturer, by fax on February 5, 1999, and March 23, 1999. Firm-initiated recall complete. DISTRIBUTION New York, Korea, Germany, Switzerland. QUANTITY 143 units were distributed. REASON Blood products tested negative for anti-HCV, but were collected from a donor who previously tested reactive for anti-HCV. ________ PRODUCT a) Red Blood Cells; b) Platelets; c) Recovered Plasma. Recall #B-073/075-0. CODE Unit #479882. MANUFACTURER Memorial Blood Centers of Minnesota, Minneapolis, Minnesota. RECALLED BY Manufacturer, by letter dated February 4, 1998. Firm-initiated recall complete. DISTRIBUTION Minnesota and Switzerland. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor who had body piercing within 12 months of donation.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS III

________ PRODUCT Source Plasma. Recall #B-010-0. CODE Unit numbers: 09111608, 09111615, 09111622, 09111639, 09111646, 09111653, 09111660, 09111677, 09111684, 09111691, 09111707, 09111714. MANUFACTURER Seramed, Inc., Tacoma, Washington. RECALLED BY Manufacturer, by telephone and fax on December 21, 1998. Firm- initiated recall complete. DISTRIBUTION California. QUANTITY 12 units were distributed. REASON Blood products tested negative for all required viral marker tests, were distributed prior to testing. ________ PRODUCT Corneas. Recall #B-011-0. CODE Unit numbers: 20837 LC and 20837 RC. MANUFACTURER Lions Eye Bank of Delaware Valley, c/o Willis Eye Hospital, Philadelphia, Pennsylvania. RECALLED BY Lions Eye Bank of Delaware Valley, Philadelphia, Pennsylvania, by telephone and fax on June 15, 1999, and by letter dated June 16, 1999. Firm-initiated recall complete. DISTRIBUTION Florida and Texas. QUANTITY 2 units were distributed. REASON Corneas were collected from a donor who tested repeatedly reactive for the HIV p-24 antigen, and non-comfirmable by PCR. ________ PRODUCT Corneas. Recall #B-012-0. CODE Unit numbers: 99-1001 and 9-1002. MANUFACTURER Georgia Eye Bank, Inc., Atlanta, Georgia. RECALLED BY Manufacturer, by telephone and letter on July 13, 1999. Firm- initiated recall complete. DISTRIBUTION Georgia and Japan. QUANTITY 2 units were distributed. REASON Corneas were collected from a donor who tested repeatedly reactive for the HIV p-24 antigen, and non-confirmable by PCR. ________ PRODUCT Source Plasma. Recall #B-017-0. CODE 30GP08319, 30GP08329, 30GP08330, 30GP08358, 30GP08368, 30GP08376, 30GP08377. MANUFACTURER American National Red Cross, Ashley, Pennsylvania. RECALLED BY Manufacturer, by telephone on April 17, 1998, and by fax on April 22, 1998. Firm-initiated recall complete. DISTRIBUTION California. QUANTITY 7 units were distributed. REASON Unlicensed blood products were distributed in interstate commerce. ________ PRODUCT a) Red Blood Cells; b) Fresh Frozen Plasma. Recall #B-024/025-0. CODE Unit #LW64240. MANUFACTURER The Blood Center of Southeastern Wisconsin, Milwaukee, Wisconsin. RECALLED BY Manufacturer, by telephone on December 16, 1998, and July 2, 1999, and by letters dated January 18, 1999, and July 6, 1999. Firm-initiated recall complete. DISTRIBUTION Wisconsin. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor taking the drug Sulfasalazine. ________ PRODUCT a) Red Blood Cells; b) Recovered Plasma; c) Red Blood Cells for Manufacturing Use; d) Platelets for Use in Manufacturing Non- Injectable Products. Recall #B-028/031-0. CODE Unit Numbers: a) 259392; b) 259392, 273070 c) 273070; d) 275600, 281147, 322010. MANUFACTURER Memorial Blood Centers of Minnesota, Minneapolis, Minnesota. RECALLED BY Manufacturer, by letter dated February 19, 1999. Firm-initiated recall complete. DISTRIBUTION Florida, Minnesota, Switzerland. QUANTITY a) 1 unit; b) 2 units; c) 1 unit; d) 3 units were distributed. REASON Blood products tested non-reactive for anti-HIV-2, but were collected from donors who previously tested repeatedly reactive for anti-HIV-2, were distributed. ________ PRODUCT Panglobulin, Immune Globulin Intravenous (Human), in 12 gram vials. Recall #B-039-0. CODE Lots 01839-00023, 01839-00012. MANUFACTURER ZLB Central Laboratory, Bern, Switzerland. RECALLED BY American Red Cross, Rosslyn, Virginia, by letter on August 20, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 7,963 vials distributed. REASON Two lots of Immune Globulin Intravenous (Human), that exceeded the residual moisture specification. ________ PRODUCT Platelets, Pheresis. Recall #B-040-0. CODE Unit #LH47927. MANUFACTURER The Blood Center of Southeastern Wisconsin, Milwaukee, Wisconsin. RECALLED BY Manufacturer, by telephone on July 7, 1999, and by letter dated July 6, 1999. Firm-initiated recall complete. DISTRIBUTION Illinois. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who had not been questioned about the use of the drug Soriatane (acitretin), ________ PRODUCT a) Whole Blood; b) Red Blood Cells, Leukocytes Reduced; c) Platelets; d) Platelets, Pooled. Recall #B-041/044-0. CODE Unit Numbers: a) LX69787; b) Q99456; c) KZ33284; d) FP96554. MANUFACTURER The Blood center of Southeastern Wisconsin, Milwaukee, Wisconsin. RECALLED BY Manufacturer, by letter dated July 6, 1999, or by telephone on November 17, 1998, December 27, 1998, and/or July 2, 1999. Firm- initiated recall complete. DISTRIBUTION Wisconsin. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from donors who had not been questioned about the use of the drug Soriatane (acitretin). ________ PRODUCT a) Red Blood Cells; b) Platelets. Recall #B-045/046-0. CODE Unit Numbers: a) 19GF42933, 19GT17811, 19GG56965, 40GL64478, 40GL64465; b) 19GL16084. MANUFACTURER American Red Cross Blood Services, Nashville, Tennessee. RECALLED BY Manufacturer, by letter dated May 13, 1999. Firm-initiated recall complete. DISTRIBUTION Tennessee. QUANTITY a) 5 units; b) 1 unit was distributed. REASON Blood products had been stored at unacceptable temperatures. ________ PRODUCT Platelets. Recall #B-063-0. CODE Unit Numbers: 53FN47150, 53T66446, 53FX71803, 53FX71794, 53T66456, 53K29270, 53FX71807. MANUFACTURER American Red Cross Blood Services, Baltimore, Maryland. RECALLED BY Manufacturer, by letter dated November 13, 1998. Firm-initiated recall complete. DISTRIBUTION Maryland. QUANTITY 7 units were distributed. REASON Blood products were shipped at an unacceptable temperature.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

________ PRODUCT Dialyzer Control: a) Centrysystem 3 Catalog Nos.: 333103-001 - US 115v Centrysystem 3, BPM 333103-101 - US 115v Centrysystem 3 Single Needle, BPM 333103-121 - Euro 240v Centrysystem 3, Single Needle, BPM 333103-201 - US 115v Centrysystem 3+, BPM 333103-301 - US 115v Centrysystem 3+, Single Needle, BPM 333104-001 - US 115v Centrysystem 3, No Options 333104-101 - US 115v Centrysystem 3, Single Needle 333104-121 - Euro 240v Centrysystem 3, Single Needle 333104-201 - US 115v Centrysystem 3+, No Options; b) Spare Transducer Assembly Catalog numbers: 501036-000 - Arterial/Venous Pressure Transducer Spare Assembly 501212-003 - Cartridge Holder Spare Assembly 501249-220 - Transducer Connector Spare Assembly. Recall #Z-064/065-0. CODE a) All devices manufactured after 11/01/1998, and all devices repaired with affected spares assemblies that were shipped after 11/01/1998. Includes machine serial numbers 3C39644 through 3C42411 (with exclusion of 460 devices within that range but released prior to 11/01/1998). b) Lot numbers with prefix 11C, 12C, 01D, 02D, 03D, 04D, 05D, 06D. MANUFACTURER Gambro Renal Care Products, Lakewood, Colorado. RECALLED BY Manufacturer, by letter on August 24, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide and international. QUANTITY 1,442 of the affected machines are in US distribution and 307 were shipped to international distributors. Also, 204 Transducer Spare Kits , 133 Cartridge Holder Spare Kits, and 3 Transducer Connector Spare Kits were distributed domestically. International distribution of these kits was 108, 8, and 0, respectively. REASON The dialyzer alarm does not sound when the Maximum Arterial Pressure Alarm Limit is challenged. ________ PRODUCT Soflex UV-Absorbing Silicone PC Intraocular Lenses Model: LI51U & LI61U. Recall #Z-066/067-0. CODE Lot Numbers 390T, 4BG1, 4BG7, 4CRA and 4DV1. MANUFACTURER Bausch and Lomb Surgical, Clearwater, Florida. RECALLED BY Manufacturer, by letter on September 14, 1999. Firm-initiated recall ongoing.. DISTRIBUTION Nationwide. QUANTITY 252 lenses were distributed; firm estimated that 124 units remained on market at time of recall initiation. REASON Intraocular lenses may exhibit a cloudy appearance. ________ PRODUCT LifeStyle MV2 Multifocal, Hydrophilic, Sterile Contact Lenses in blister packs, Rx product. Recall #Z-073-0. CODE All lots. MANUFACTURER St. Shine Optical Company, Ltd., Taiwan, Republic of China. RECALLED BY The LifeStyle Company, Inc., Morganville, New Jersey, by letter dated August 30, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY Approximately 147,840 units were distributed. REASON The contact lenses contain sorbic acid that is not declared on the labeling and may cause a burning sensation in the patient's eye. ________ PRODUCT Locking Blot Measuring Device, used to determine the required length of the locking bolt: a) Part number 357.792; b) and Part number 357.113.311. Recall #Z-075/076-0. CODE Lot numbers: a) A4GG415; b) lot A4GF199. MANUFACTURER Synthes USA, West Chester, Pennsylvania. RECALLED BY Synthes USA, Paoli, Pennsylvania, by E-mail on November 7, 1997. Firm-initiated recall complete. DISTRIBUTION Arkansas, Florida, Minnesota, North Carolina, New Jersey, Pennsylvania, Germany. QUANTITY 22 units were distributed. REASON The ball that attaches to the measuring slider of the device may detach. ________ PRODUCT Synthes Anatomical Locking Plate System, used to treat fractures of long bones such as the femur: a) Part number 422.340.111; b) Part number 422.341.111; c) Part number 422.344.111; d) Part number 422.345.111; e) Part number 422.348.111 and f) Part number 422.349.111. Recall #Z-078/083-0. CODE Lot numbers: a) N2002, N2002.1, N2003, N2004, N2005; b) N2003, N2003.1, N2004, N2005; c) N2001, N2001.1, N2002, N2003, N2004, and N2005; d) N2001, N2001.1, N2002, N2003, N2004, N2005; e) N2001, N2001.1, N2002, N2003, N2004, N2005; f) N2001, N2001.1, N2002, N2003, N2004, N2005. MANUFACTURER Mathys Medical Ltd., Bettlach, Switzerland. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on December 19, 1997. Firm-initiated recall complete. DISTRIBUTION Alabama, Colorado, Florida, Mississippi, North Carolina, New York. QUANTITY 288 units were distributed. REASON The fatigue strength of the plates did not meet the original specification. ________ PRODUCT Proximal Femoral Nail (PFN) System, used to ream a hole for the insertion of a femoral neck screw. Recall #Z-084-0. CODE Part #357.044, Lot Numbers 1006 through 1015. MANUFACTURER Stratec Medical, Eimattstrasse, Switzerland. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on September 2, 1997, and December 1, 1997. Firm-initiated recall complete. DISTRIBUTION California, Connecticut, Illinois, Pennsylvania, Canada. QUANTITY 14 units were distributed. REASON The tip of the device may break when it comes in contact with a bent guidewire.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III

________ PRODUCT Navarre 6 French J-Curve Universal Drainage Catheter, Product reference number NUD6J, used only in non-vascular drainage procedures, and are intended to be used for abscess, cyst, and other general purpose drainage applications. Recall #Z-023-0. CODE Lot 990610, Use by 2002-04. MANUFACTURER Navarre Biomedical, Ltd., Plymouth, Minnesota. RECALLED BY Manufacturer, by memorandum on September 17, 1999. Firm- initiated recall complete. DISTRIBUTION Georgia, Indiana, Pennsylvania, Australia, Canada. QUANTITY 105 catheters were distributed. REASON The catheter size was incorrectly identified at the tops of the pouch labels of the catheters as 8 French (inside diameter), rather than as the actual size (6 French). ________ PRODUCT Pelvic Implant Set Graphic Case for Self-Tapping Screws, used to organize and store implants including tapping screws. Recall #Z- 070-0. CODE Part #306.20, Lot #VC2671. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on November 25, 1997. Firm-initiated recall complete. DISTRIBUTION California, Georgia, Iowa, South Carolina. QUANTITY 4 units were distributed. REASON Mislabeling - The rack in the graphic case that hold the self tapping screws is labeled with the part numbers for the standard screws instead of the self tapping screws. ________ PRODUCT Synthes Mandibular Modular Fixation System. Recall #Z-074-0. CODE Part number 449.014, Lot A3KJ682. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on October 18,1997. Firm-initiated recall complete. DISTRIBUTION Alabama, California, Colorado, Florida, Hawaii, Michigan, Missouri, North Carolina, New York, Ohio, Pennsylvania, South Carolina, Texas, Washington, Wisconsin. QUANTITY 43 units were distributed. REASON One of the outer labels of the device was purple instead of blue. A purple label corresponds with 2.4-mm plates. The color of the label is used to indicate the diameter of the screw that is used with the plate. The label still read what type and length the plate was in the package. The user would have realized that the color was incorrect. ________ PRODUCT Pelvic Implant Set Graphic Case for Self-Tapping Screws. Recall #Z-077-0. CODE Part #306.20, Lot VC2671. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on November 25, 1997. Firm-initiated recall complete. DISTRIBUTION California, Georgia, Iowa, South Carolina. QUANTITY 4 units were distributed. REASON Mislabeling - The rack in the graphic case that hold the self tapping screws is labeled with the part numbers for the standard screws instead of the self tapping screws. ________ PRODUCT Synthes TI-6AI-7NB Unreamed Femoral Nail, used for intramedullary stabilization for femoral shaft and subtrochanteric fractures, ipsilateral femoral neck and shaft fractures and impending pathologic fractures of the femur.. Recall #Z-087-0. CODE Part number 464.34S, Lot A3CX522. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by letter sent on April 8, 1998. Firm-initiated recall complete. DISTRIBUTION Arizona, California, Colorado, Florida, Massachusetts, Missouri, North Carolina, Nebraska, New Jersey, New York, Ohio, Pennsylvania. QUANTITY 6 units were distributed. REASON Mislabeling - The package was labeled as "9 x 340 mm Titanium Unreamed Femoral Nail" but contained an "11 x 340 mm Titanium Unreamed Femoral Nail". ________ PRODUCT Synthes Unreamed Humeral Nail (URHN). Recall #Z-088-0. CODE Part number 458.40S, Lot #A3JL013. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on September 19, 1997, followed by visit. Firm-initiated recall complete. DISTRIBUTION Nationwide and Canada. QUANTITY 95 units were distributed. REASON The locking bolts were packaged as 40 mm in length but were actually 42mm in length. ________ PRODUCT Bone Staple(s): a) Part #282.80; b) Part #282.82. Recall #Z-089/090-0. CODE Lot Numbers: a) 2022, 2031, 2032, 2033, and 2034; b) 2026. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on October 10, 1997. Firm-initiated recall complete. DISTRIBUTION California, Kansas, Missouri, North Carolina, New York, Ohio, Oklahoma, Pennsylvania, Texas. QUANTITY 250 units were distributed. REASON Mislabeling - The package label states "intended for femoral fractures and osteotomies" which is not applicable for this device. ________ PRODUCT Saw Guide Assembly, Right and Left, HTO System, used as a guide for a bone saw: a) Part #P94025.013111; b) Part P94025.014011. Recall #Z-091/092-0. CODE Lot numbers: a) A4GF675; b) A4GF681. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by fax and telephone on September 22, 1997. Firm-initiated recall complete. DISTRIBUTION Missouri, Mississippi, Germany. QUANTITY 6 units were distributed. REASON The stud that is used to secure the saw guide to the guidewire may loosen from the base. ______ UPDATE The following two recalls which appeared in the October 6, 1999 Enforcement Report were re-classified from Class II to Class III recalls and should read as follows: PRODUCT Model 35040 Keithley Therapy Dosimeter, intended use for calibration of dosimetry of therapeutic radiation treatment machine for high-energy accelerators, cobalt 60, and brachytherapy equipment. Recall #Z-1261-9. CODE Serial Numbers: 69450-69469; 80276-80295; 82666-82685; 81909-81928; and 86087-86106. MANUFACTURER Inovision Radiation Measurements, Cleveland, Ohio. RECALLED BY Manufacturer, by letters on July 8, 1999, and September 10, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide and international. QUANTITY 89 units. REASON A 1.6 Amp fuse may have been installed where a 1 Amp is specified. ________ PRODUCT The Tracker Display Model #35360A is sold with the Detector Model #35300A and marketed together as the Keithley Model #90100 Tracker System, a radiation measurement system intended for use in quality assurance programs for high energy accelerators, and cobalt 60 machines. Recall #Z-1262-9. CODE Serial Numbers: 83728-83747 and 84582-84601. MANUFACTURER Inovision Radiation Measurements, Cleveland, Ohio. RECALLED BY Manufacturer, by letters on July 8, 1999, and September 10, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide and international. QUANTITY 40 units. REASON A 1.6 Amp fuse may have been installed where a 1 Amp is specified.

RECALLS AND FIELD CORRECTIONS: VETERINARY PRODUCTS -- CLASS III

________ PRODUCT Consolidated Nutrition Metabalance 15/25 and Consolidated Nutrition Metabalance 25/45, in 50 pound bags, containing 100 gm/ton chlortetracycline, 0.011% sulfamethazine and 50 gm/ton penicillin - complete feed for starter pigs Recall #V-038-0. CODE Lot Nos - Metabalance 15/25 - GLO4799 Metabalance 25/45 - GLO429 - 50 lb bag GLO489 - bulk GLO609 - 50 lb. bag GLO639 - bulk GLO699 - bulk GLO689 - 50 lb bag GLO839 - bulk GLO909 - bulk (2 batches). MANUFACTURER Consolidated Nutrition, L.C., Glencoe, Minnesota. RECALLED BY Consolidated Nutrition, L.C., Omaha, Nebraska, by telephone on or about July 14, 1999. Firm-initiated recall complete. DISTRIBUTION Minnesota. QUANTITY 78 50-lb bags of Metabalance 15/25, 484 50-lb bags and 61,800 pounds of bulk of Metabalance 25/45 were distributed. REASON Products were manufactured with incorrect Type A Medicated Article - Aureozol 500 containing sulfathiazole. ________ PRODUCT a) Bulk Consolidated Nutrition Commercial 25/50 Medicated containing 100 gm/ton chlortetracycline, 0.011% sulfamethazine and 50 gm/ton penicillin complete feed for starter pigs - b) Bulk Consolidated Nutrition Metabalance 25/45 containing 100 gm/ton chlortetracycline, 0.011% sulfamethazine and 50 gm ton penicillin complete feed for starter pigs-. c) Alliance Animal Health Super 250 8463 Medicated, containing 10 gm/lb chlortetracycline, 2.2% sulfamethazine and 5 gm/lb penicillin Type B premix for cervical abscesses, bacterial swine enteritis, prevention of these diseases during stress, weight gain and increased feed efficiency d) Alliance Animal Health Advanced Stress Formula Aureo SP-250 Option 8541 Medicated, containing 4 gm/lb chlortetracycline, 0.88% sulfamethazine and 2 gm/lb penicillin, type B premix - treatment of cervical abscesses, bacterial swine enteritis, prevention of these diseases during stress and maintenance of weight gains and growth promotion. Recall #V-039/042-0. CODE Lot Numbers: a) LM 3418, LM3518, LM3638, LM0199, LM0399, LM0689, LM0849, LM0919, LM0989, LM1189, LM1319, LM1659, LM1799; b) LM0089 and LM1379; c) LM1529; d) LM3388 and LM0289. MANUFACTURER Consolidated Nutrition, L.C., LeMars, Iowa. RECALLED BY Consolidated Nutrition, Omaha, Nebraska, by telephone on July 14, 1999. Firm-initiated recall complete. DISTRIBUTION a) South Dakota; b) South Dakota; c) Iowa and Indiana; d) Indiana, South Dakota, Nebraska, Iowa. QUANTITY a) 129,620 pounds; b) 27,680 pounds; c) 155 20-pound bags; d) 236 25-pound bags were distributed. REASON Products were manufactured using incorrect Type A Medicated Article, Aureozol 500 containing sulfathiazole. ________ PRODUCT Consolidated Nutrition Metabalance 25/45 10485AXU Medicated containing 100 gm/ton chlortetracycline, 0.011% sulfamethazine and 50 gm/ton penicillin completed feed for starter pigs. Recall #V-043-0. CODE Lot Numbers: NA0999, NA1199, NA1319, NA1389, and NA1449. MANUFACTURER Consolidated Nutrition, L.C., Nappanee, Indiana. RECALLED BY Consolidated Nutrition, L.C., Omaha, Nebraska, by telephone on July 14, 1999. Firm-initiated recall complete. DISTRIBUTION Michigan, Ohio, Indiana. QUANTITY 555 50-pound bags were distributed. REASON Product was manufactured with incorrect Type A Medicated Article, Aureozol 500 containing sulfathiazole. ________ PRODUCT Aureozol 500 Granular Type A Medicated Article - containing 40lb/chlortetracycline calcium complex equivalent to chlortetracycline HCl, 8.8%, 40 g lb sulfathiazole, 20 g lb. Penicillin Recall #V-044-0. CODE Lot D821-3. MANUFACTURER Consolidated Nutrition, Omaha, Nebraska. RECALLED BY Manufacturer, by telephone on or about July 14, 1999. Firm-initiated recall complete. DISTRIBUTION Illinois, Minnesota, Indiana, Iowa. QUANTITY 48 50-pound bags were distributed. REASON Aureozol 500 shipped to 4 plants under incorrect ingredient number 70192. Aureomix 500. Feed manufactured from this shipment contained the incorrect sulfa drug. END OF ENFORCEMENT REPORT FOR NOVEMBER 3, 1999. ####