FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contain s information on actions taken in connection with agency regulatory activities.
 ENFORCE 
09/17/1996

 ENFORCEMENT REPORT FOR 09/18/96 

September 18, 1996                              96-38

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Frozen cookie dough pellets and frozen cheese
               cake chunks extruded and cut to dimensions and
               utilized as an add-in for further processing
               into ice cream/yogurt/ice milk products,
               fudge, etc.: 
               (a) AF-01-500 Non Coated Chocolate Chip Add
               In, in 20 pound boxes
               (b) AF-01-501 Coated Chocolate Chip Add In, in
               20 pound boxes
               (c) AF-01-505 Coated Low Fat Fudge Brownie Bit
               Add In, in 20 pound boxes
               (d) AF-01-509 Coated Premium No Chip Add In,
               in 40 pound boxes
               (e) AF-01-511 Coated Chocolate Chip Add In, in
               40 pound boxes
               (f) AF-01-517 Coated Chocolate Chip Natural
               Add In, in 40 pound boxes
               (g) AF-01-550  Coated Fudge Brownie Bit Add
               In, in 40 pound boxes.  Recall #F-694/700-6.
CODE           All misbranded product.
MANUFACTURER   Rhino Foods, Inc., Burlington, Vermont.
RECALLED BY    Manufacturer, by telephone on July 10, 1996,
               followed by letter dated July 11, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Connecticut, Massachusetts, New Hampshire, New
               Jersey, New York, Ohio, Pennsylvania, Vermont,
               Wisconsin, District of Columbia.
QUANTITY       Estimate of Inventory in the Field:
               AF-01-500 - 163 cases
               AF-01-501 - 382 cases
               AF-01-505 - 737 cases
               AF-01-509 - 353 cases
               AF-01-511 - 215 cases
               AF-01-517 - 309 cases
               AF-01-550 - 571 cases.
REASON         Products were shipped without ingredient
               statements and contained liquid pasteurized
               eggs. The products contain undeclared eggs.

_______________
PRODUCT        Genesee Farms Cookie Dough Ice Cream, in 3-
               gallon tubs.  Recall #F-708-6.
CODE           6184.
MANUFACTURER   Genesee Farms, Attica, New York.
RECALLED BY    Genesee Farms, Batavia, New York, by letters
               on August 9, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   New York.
QUANTITY       76 tubs were distributed.
REASON         Product contains undeclared eggs.

_______________
PRODUCT        Kellogg's Frosted Bran Cereal, 18.9 ounce box. 
               Recall #F-717-6.
CODE           "Better if used before" date of "APR 26 1997
               NA 009" printed on the boxtop.
MANUFACTURER   Kellogg Company, Omaha, Nebraska.  
RECALLED BY    Kellogg Company, Battle Creek, Michigan, by
               press release on August 13, 1996, and by visit
               beginning on August 14, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Alabama, Connecticut, Florida, Georgia,
               Indiana, Kentucky, Massachusetts, Maine,
               Michigan, Mississippi, New Jersey, New York,
               North Carolina, Ohio, Pennsylvania, South
               Carolina, Tennessee, Virginia, West Virginia.
QUANTITY       Approximately 4,400 cases (12 boxes per case)
               were distributed.
REASON         Product contains undeclared peanuts.


                             -2-RECALLS AND FIELD CORRECTIONS:  CLASS II ====================
_______________
PRODUCT        Frozen cookie dough pellets and frozen cheese
               cake chunks extruded and cut to dimensions and
               utilized as an add-in for further processing
               into ice cream/yogurt/ice milk products,
               fudge, etc.: 
               (a) AF-01-507  No Fat Coated Chocolate Chip
               Add In, in 20 pound boxes
               (b) AF-01-514  Coated No Egg No Chip Add In,
               in 40 pound boxes, in 40 pound boxes
               (c) AF-01-603  Coated No Egg Chocolate Chip
               Coconut Add In, 40 pound boxes
               (d) AF-01-605  Coated No Egg Chocolate Chip
               Add In, in 40 pound boxes
               (e) AF-01-700  Coated Cheesecake Add In, in 20
               pound boxes.  Recall #F-701/705-6.
CODE           All misbranded product.
MANUFACTURER   Rhino Foods, Inc., Burlington, Vermont.
RECALLED BY    Manufacturer, by telephone on July 10, 1996,
               followed by letter dated July 12, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Connecticut, Massachusetts, New Hampshire, New
               Jersey, New York, Ohio, Pennsylvania, Vermont,
               Wisconsin, District of Columbia.
QUANTITY       Estimate of inventory in the field:
               AF-01-507 - 248 cases
               AF-01-514 - 204 cases
               AF-01-603 -  30 cases
               AF-01-605 - 430 cases
               AF-01-700 -  26 cases.
REASON         Products were shipped without ingredient
               statements and contained undeclared wheat
               flour.

_______________
PRODUCT        Lay's Regular Potato Chips, in 1.5 ounce vend
               size bags.  Recall #F-713-6.
CODE           OCT 15 434206 1.5 ounce vend size Lay's
               regular potato chips.
MANUFACTURER   Frito-Lay, Inc., Louisville, Kentucky.
RECALLED BY    Frito-Lay, Inc., Plano, Texas, by telephone
               beginning August 12, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Indiana.
QUANTITY       84 cases (36 bags per case) were distributed.
REASON         Product contains undeclared dairy ingredients
               and bears an inappropriate identity statement.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
                             -3-_______________
PRODUCT        Zia Maria's Italian Spaghetti Sauce, in 16
               ounce glass jars with one-piece metal screw-on
               lids.  Recall #F-691-6.
CODE           Uncoded.
MANUFACTURER   Palmieri Food Products, Inc., New Haven,
               Connecticut.
RECALLED BY    Zia Maria's, Inc., Minden, Nebraska, by
               telephone on June 4, 1996, and by television
               broadcast on June 4 & 5, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   Nebraska.
QUANTITY       14 cases (12 jars per case) were distributed.
REASON         Some jars have popped up lids and are spoiled.

_______________
PRODUCT        Tortellini with Cheese Filling Plain and
               Tortellini with Cheese Filling Tricolor, in 8
               ounce bags and 2.2 pound bags.  
               Recall #F-692/693-6.
CODE           All codes.
MANUFACTURER   Audisio Ind. Alim. S.R.L., Italy.
RECALLED BY    California Specialty Food Importers, Inc.,
               Chatsworth, California (importer), by
               telephone, August 9, 1996, followed by letter
               sent August 13, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   California, Nevada, Washington state.
QUANTITY       38 cases (6 2.2-pound bags and 22 cases (24 8-
               ounce bags per case) were distributed.
REASON         Product does not list a complete ingredient
               statement and contains undeclared hot pressed
               peanut oil.

_______________
PRODUCT        Various Ocean Spray Cranberry Juices, in 1
               gallon plastic bottles:  (a) Ocean Spray
               Cranberry Juice Cocktail; (b) Ocean Spray
               Cranapple Cranberry Apple Juice Drink; (c)
               Ocean Spray Cran-Raspberry Raspberry Cranberry
               Juice Drink; (d) Ocean Spray Ruby Red
               Grapefruit Juice Drink.  Recall #F-709/712-6.
CODE           Cap Codes:  (a) 03037K....LTPCJ2 and
               03047K....LTPCJ2; (b) 03047K....LTPCA2; (c)
               03057K....LTPCR2; (d) 03057K....LTPRR2 &
               03067K....LTPRR2.
MANUFACTURER   Ocean Spray Cranberries, Inc., Kenosha,
               Wisconsin.
RECALLED BY    Ocean Spray Cranberries, Inc., Lakeville-
               Middleboro, Massachusetts, by memorandum on
               August 22, 1996, followed by visit.  Firm-
               initiated recall ongoing.

                             -4-DISTRIBUTION   Nationwide.
QUANTITY       20,913 cases (6 bottles per case) were
               distributed.
REASON         Product is contaminated with mold.

_______________
PRODUCT        Aloha Maid iced tea, lemonade, and fruit
               drink, in 20 fluid ounce plastic bottles: (a)
               Iced Tea; (b) Strawberry Flavored Lemonade;
               (c) Orange Passion Fruit Drink.  
               Recall #F-714/716-6.
CODE           (a) Open date code: 960605A, B or C, and
               960607B; (b) 960606E, F, or G, and 960607A;
               (c) 960606A, B, C, or D.
MANUFACTURER   Cliffstar Corporation, Fontana, California.
RECALLED BY    Itoen (USA), Inc., Honolulu, Hawaii, by visit
               and by telephone beginning on August 14, 1996,
               and by press release on August 16, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Hawaii and Guam.
QUANTITY       Firm estimates none of the May or June 1996
               production remains on market.
REASON         Product is unfit for food based on off-taste,
               alcohol-like taste and leaks.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Thiothixene HCl Capsules, USP, 1 mg, 2 mg and
               5 mg strength, packed in bottles of 100, under
               the Schein and Rugby labels, an antipsychotic. 
               Recall #D-263/265-6.
CODE           1 mg, Lot numbers:  C5C0227, CXB104, CXC104
               2 mg, Lot numbers:  C5C0231
               5 mg, Lot numbers:  C5E0809, C5E1128, C5E1129
               (Schein label)
               1 mg, Lot numbers:  CXB104, CXC104
               2 mg, Lot numbers:  C5C0231 (Rugby label).
MANUFACTURER   Danbury Pharmacal, Inc., Carmel, New York.
RECALLED BY    Danbury Pharmacal, Inc., Subsidiary of Schein
               Pharmaceutical, Inc., Brewster, New York, by
               letter dated July 31, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       28,710 bottles (1 mg); 9,513 bottles (2 mg);
               26,752 bottles (5 mg) were distributed.
REASON         Dissolution failure.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III =====
_______________
PRODUCT        Red Blood Cells.  Recall #B-625-6.
CODE           Unit numbers:  L50665, L50666.
                             -5-MANUFACTURER   South Bend Medical Foundation, Inc., South
               Bend, Indiana.
RECALLED BY    Manufacturer, by telephone on or about July
               21, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Indiana.
QUANTITY       2 units were distributed.
REASON         Blood products were mislabeled with an
               incorrect expiration date.

_______________
PRODUCT        Red Blood Cells.  Recall #B-626-6.
CODE           Unit numbers:  38J35701, 38J35702, 38J35704,
               38J35705, 38J35706, 38J35707.
MANUFACTURER   American Red Cross, Fort Wayne, Indiana.
RECALLED BY    Manufacturer, by telephone on April 13, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Ohio, Tennessee, Indiana.
QUANTITY       6 units were distributed.
REASON         Blood products were mislabeled with an
               incorrect expiration date.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Mobile/Portable General Purpose Units, used in
               radiographic studies:
               (a) Model HP200; (b) HP300.
               Recall #Z-1101/1102-6.
CODE           HP200 and HP300 mobile/portable general
               purpose systems containing the R-120 series
               collimators.
MANUFACTURER   Mikasa X-Ray Co LTD., Tokyo, Japan.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan on September 9, 1996. 
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       40 units.
REASON         Noncompliance with performance standards for
               diagnostic x-ray products in that the units
               failed to meet light illuminance requirement
               at a source to image distance (SID) of 40
               inches and they failed to meet edge contrast
               ratio requirement at an SID of 40 inches.

_______________
PRODUCT        Transmucosal Implant Extensions, endosseous
               implant accessories used for dental
               restoration surgery:
               (a) Second Phase Set, Catalog No. 8246S;  4.0
               mm Diameter x 4.0 mm High;
               (b) Transmucosal Implant Extension (TIE)
               Catalog No. 8265A,  4.0 mm Diameter x 4.0 mm
               High.  Recall #Z-1133/1134-6.
                             -6- CODE          Lot Numbers:  (a) 302406, 307908, 301311,
               413002, 413704, 409106, 421206, 412908,
               407809, 417209, 404210, & 416510; (b) 305909,
               412909, 413503, 405704, 400905, 407606,
               422006, 407606, 412508.
MANUFACTURER   Interpore International, Irvine, California.
RECALLED BY    Manufacturer, by letters on January 31, 1995.
               and March 10, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide, Mexico, Lebanon, Canada, Taiwan.
QUANTITY       1,926 units were distributed; firm estimated
               that little if any product remained on market
               at time of recall initiation.
REASON         The packaging labels were labeled as 4.0 x 1.0
               mm IMC Titanium Inserts when the packages
               actually contained products with a 4.0 x 4.0
               mm Transmucosal Implant Extension.

_______________
PRODUCT        Bactec Instrument 460 TB Drug Susceptibility
               Testing Instructional Manual.  
               Recall #Z-1144-6.
CODE           All lot numbers.
MANUFACTURER   Becton Dickinson Diagnostic Instrument
               Systems, Sparks, Maryland.
RECALLED BY    Manufacturer, by Important Technical Bulletin. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Puerto Rico.
QUANTITY       657 manuals were distributed.
REASON         The manual for the device contains inadequate
               instructions of use.

_______________
UPDATE         Davis-Geck Valtrac Bar Biofragmentable
               Anastomosis Ring, Recall #Z-1043/1051-6 which
               appeared in the August 28, 1996, Enforcement
               Report listed an incorrect product number for
               item (b).  The correct product number is 
               8083-00.
               RECALLED BY also failed to provide the
               following information:  Sherwood Medical
               Company, St. Louis, Missouri, issued two
               letters dated May 22, 1996, to U.S. accounts
               and a letter dated May 23, 1996, to foreign
               accounts, requesting return of only Control
               No. 894100 of Product No. 8083-00, and that
               the consignees examine all other lot numbers
               (Control Nos.) of all Product Numbers listed
               for proper orientation of the device.  Any
               device found not properly oriented was not to
               be used, but returned for replacement.  Firm-
               initiated field correction ongoing.

                             -7-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        Human Albumin, 30% Solution Nonsterile,
               Catalog #3110, for use in laboratory research
               or manufacturing non-injectable products only. 
               Recall #Z-984-6.
CODE           Lot #4001657.
MANUFACTURER   Biocell Laboratories, Inc., Rancho Dominguez,
               California.
RECALLED BY    Manufacturer, by telephone on June 11, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       40 50-ml vials were distributed; firm
               estimates none remains on the market.
REASON         The device fails to meet it's labeling
               specification of 3 ml.  It was determined that
               the device was bottled 2 ml short.

_______________
PRODUCT        Hip Stem Endoprosthesis:
               (a) Sentry, Catalog No. 6258-5-008, 5 Right,
               250mm Length, 15.5mm Diameter;
               (b) Sentry, Catalog No. 6258-5-005, 4 Right,
               250mm Length, 15.5mm Diameter;
               (c) Sentry, Catalog No. 6258-4-012, 5 Left,
               250mm Length, 21.5mm Diameter;
               (d) Austin Moore, Catalog No. 6940-9-400, 
               Labeled 9" in Length.  Recall #Z-1127/1130-6.
CODE           Lot Codes:  (a) ATMNB; (b) CHXZA; (c) CHTZA;
               (d) OKYVA.
MANUFACTURER   Howmedica, Inc., Rutherford, New Jersey.
RECALLED BY    Manufacturer, (a) by fax on November 17, 1995;
               (b) by fax on January 10, 1996; (d) by fax on
               December 7, 1995. Firm-initiated recall
               complete.
DISTRIBUTION   (a) Massachusetts, New Jersey, Virginia; (b)
               Minnesota, California, (c) Virginia; (d)
               Minnesota.
QUANTITY       Amount shipped:  (a) Lot code: ATMNB - 4
               units, CCKFA - 0 units, (b) Lot code: CHXZA -
               2 units; (c) Lot code: CHTZA - 1 unit; (d) Lot
               code: OKYVA - 2 units.
REASON         Packaging, labeling, and sizing mix-ups led to
               packages with incorrect labeling or product.

_______________
PRODUCT        Head and Neck Replacement (HNR) Log Stem,
               Catalog #6257-2-300.  Recall #Z-1131-6.
CODE           Lot Code COGSC.
MANUFACTURER   Howmedica, Inc., Rutherford, New Jersey.
RECALLED BY    Manufacturer, by fax on February 21, 1996. 
               Firm-initiated recall complete.

                             -8-DISTRIBUTION   Pennsylvania, West Virginia, Georgia,
               Michigan, Washington state, Ohio.
QUANTITY       8 units were distributed.
REASON         Instead of containing a 200 mm x 43 mm long
               stem as labeled, the package may contain a 225
               mm x 43 mm long stem.

_______________
PRODUCT        Package insert associated with Dade
               Immunoassay Control Comprehensive Tri-Level
               Kit, in-vitro diagnostic kit used to assist in
               monitoring accuracy and precision in clinical
               assays:  (a) Catalog Nos. B5700-05, B5700-S;
               (b) Catalog No. B5700-06; (c) Catalog No.
               B5700-07; (d) Catalog No. B5700-08.  
               Recall #Z-1135/1138-6. 
 CODE          Lot Numbers:  (a) IACK-28; (b) IAC1-128; 
               (c) IAC2-228; (d) IAC3-328.  All kits expire
               on March 26, 1998.
MANUFACTURER   Dade International, Inc., Miami, Florida.
RECALLED BY    Manufacturer, by letter dated June 1996
               containing corrected package insert.  Firm-
               initiated field correction complete.
DISTRIBUTION   Nationwide and Italy.
QUANTITY       10,776 kits were distributed.
REASON         The package insert had multiple
               inconsistencies that included an incorrect
               assay range.

_______________
PRODUCT        Incstar FTA-ABS Flouro Test Kit, in-vitro
               diagnostic for detecting antibodies in human
               serum to syphilis: (a) - Catalog No. 1910; (b)
               Catalog No. 1910G; (c) Catalog No. 6910; (d)
               Incstar FTA-ABS Test Sorbent, Catalog No.
               1866.  Recall #Z-1139/1142-6.
CODE           Lot Numbers:  (a) 115516, 115606A, 175436,
               175446, 195796, 215126, 195796B, 142016,
               142016A, 205356, 205356A, 265356, 265376,
               295196, 265366;
               (b) 142016, 142016A, 205356, 205356A;
               (c) 115506, 175426A, 175436A, 175416, 195796A,
               215216A, 265376A, 295196A;
               (d) 122156, 122156A, 122156B.
MANUFACTURER   Incstar Corporation, Stillwater, Minnesota.
RECALLED BY    Manufacturer, by letter on August 12, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       Incstar FTA-ABS Fluoro Test Kits, product
               numbers 1910, 1910G and 6910: 587 of the test
               kits were distributed.  Incstar FTA-ABS Test
               Sorbent, product number 1866: 30 of the test
               kits were distributed.
                             -9-REASON         Some negative samples may yield a positive
               assay response.

_______________
PRODUCT        C. Albicans Growth Promotion Test Suspension,
               Catalog #GP-03, used for USP growth promotion
               testing.  Recall #Z-1149-6.
CODE           Lot numbers:  GCA010-010, GCA010-012.
MANUFACTURER   North American Science Associates (NAMSA),
               Northwood, Ohio.
RECALLED BY    Manufacturer, by letter August 28, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Ireland, Japan, Singapore.
QUANTITY       110 vials were distributed.
REASON         There was a reduction in spore population
               below the labeled range.

_______________
PRODUCT        EP Medical Electrophysiology Catheters 2mm and
               2, 5, 2mm electrode spacing intended for ECG
               recording, intercardiac pacing, stimulation
               and sensing in sterile form:
               (a) Model EPM-64-SC-252; (b) Model
               EPM-64-HD-2; (c) Model EPM-64-SC-2.
               Recall #Z-1150/1152-6.
CODE           Lot #95117, use before 6/97. 
MANUFACTURER   ProCath Corporation Subsidiary of EP Medical,
               Inc., Berlin, New Jersey.
RECALLED BY    Manufacturer, by letter on October 17, 1995,
               and January 19, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       21 catheters were distributed.
REASON         The tubing weld joints may separate if handled
               forcefully prior to insertion.  The separation
               of the tubing weld joint can result in a loss
               of torque of the distal tip curve.  A loss of
               torque would render the catheter ineffective.

_______________
PRODUCT        VAI Cultureset Herpes Simplex Virus (HSV)
               Identification Kit 501020, Immunoperoxidase
               Staining Kit for use with Viral Tissue Culture
               Isolation, for in-vitro diagnostic use. 
               Recall #Z-1145-6.
CODE           Lot 106, EXP 26 MAY 97.
MANUFACTURER   VAI Diagnostic, Memphis, Tennessee.
RECALLED BY    Manufacturer, by telephone or letter on June
               21, 1996.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       224 kits were distributed.

                            -10-REASON         The device has exhibited an unexpected drop in
               the reactivity of the labeling reagent,
               included in the kit.  As a result, the
               intensity of staining in positive samples may
               reduced.

_______________
PRODUCT        Osteonics Omnifit Cup Insert series, sterile,
               implantable, non-critical Rx devices:
               (a) Osteonic Omnifit 10 Degree cup Insert
               Series I, 28 mm, Catalog #2002-2854; (b)
               Osteonics Omnifit 20 Degree Cup Insert Series
               II, 26 mm, Catalog #2042-2654.  
               Recall #Z-1153/1154-6.
CODE           (a) Case # 1NCGW; (b) Case #1NCGM.
MANUFACTURER   Osteonics Corporation, Allendale, New Jersey.
RECALLED BY    Manufacturer, by telephone on April 22, 1996,
               followed by letter on April 30, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 24 units; (b) 11 units were distributed.
REASON         The 20 degree labeled package may contain the
               10 degree unit, and vice-versa.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
============
_______________
PRODUCT        Gentamicin Sulfate Solution, 100 mg/ml, for
               the control of bacterial infections of the
               uterus in horses and as an aid in improving
               conception in mares with uterine infections
               caused by bacteria sensitive to gentamicin.
               Recall #V-039-6.
CODE           Lot #51618 EXP JAN 98.
MANUFACTURER   Rhone Merieux, Inc., Fort Dodge, Iowa.
RECALLED BY    Rhone Merieux, Inc., Athens, Georgia, by
               letter dated July 25, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   Missouri.
QUANTITY       11,040 250-ml vials were distributed.
REASON         Precipitate in the product.

                            -11-


END OF ENFORCEMENT REPORT FOR SEPTEMBER 18, 1996.  BLANK PAGES MAY FOLLOW.

                                   ####