FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contain s information on actions taken in connection with agency regulatory activities.
 ENFORCE 
09/03/1996

 ENFORCEMENT REPORT FOR 09/04/96 

September 4, 1996                                              96-36

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I =========================

_______________
PRODUCT:       (a) Vanilla Cookie Dough Fudge packaged in 5 ounce and 9 ounce
               packages; (b) Double Chocolate Brownie Fudge & Vanilla Cookie
               Dough Fudge multi package under the Hebert Candies label. 
               Recall #F-679/680-6.
CODE:          Product manufactured on June 26, 1996 and distributed between
               June 26, 1996 through July 10, 1996.
MANUFACTURER:  Home of Hebert Candies, Inc., Shrewsbury, Massachusetts.
RECALLED BY:   Manufacturer by telephone July 10, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION:  Massachusetts, Maine, New Hampshire.
QUANTITY:      Approximately 153 pounds were distributed.
REASON:        The product contain undeclared liquid pasteurized eggs.

_______________
PRODUCT:       Smoked trout, Recall #F-682-6.
CODE:          "4510".
MANUFACTURER:  Shuckman's Food Company, Louisville, Kentucky.
RECALLED BY:   Manufacturer by telephone on July 18, 1991.  Firm-initiated
               recall complete.
DISTRIBUTION:  Kentucky.
QUANTITY:      30 pounds were distributed.
REASON:        The product is contaminated with Listeria monocytogenes.
_______________
PRODUCT:       Mercer's brand Cookie Dough Ice Cream in half gallon and 2 1/2
               gallon tubs, Recall #F-683-6.
CODE:          Not coded.
MANUFACTURER:  Mercer's Dairy Inc., Booneville, New York.
RECALLED BY:   Manufacturer by letter on or about July 31, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION:  New York.
QUANTITY:      1,930 half gallons were distributed.
REASON:        The product contains undeclared eggs.

_______________
PRODUCT:       Chocolate Chip Cookie Dough Ice Cream in half gallons and
               square pints, Recall #F-684-6.
CODE:          5A17, 5K07, 6D11, 6D24, 6E23, 6P17, 6P27 (1/2 gallon); 5A16,
               6D12, 6E01, 6P01, 6017, 6S10.
MANUFACTURER:  Hersey Creamery Company, Harrisburg, Pennsylvania.
RECALLED BY:   Manufacturer by telephone June 27, 1996, and press release on
               June 28, 1996.  Firm-initiated recall ongoing.
DISTRIBUTION:  Midwest and Eastern United States.
QUANTITY:      42,834/half gallon; 51,184 pints were distributed.
REASON:        The product contains undeclared eggs.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II =========================

_______________
PRODUCT:       Strawberry Cheese Cake Ice Cream, bulk 3 gallon containers,
               1/2 gallon containers, quarts and pints, Recall #F-681-6.
CODE:          E23, F17, G5.
MANUFACTURER:  University of Wisconsin-Madison, Madison, Wisconsin.
RECALLED BY:   Manufacturer by letter July 30, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION:  Wisconsin.
QUANTITY:      141/3 gallon containers; 70/half gallon containers; 115 
               quarts; 40 pints were distributed.
REASON:        The product does not list a complete ingredient statement and
               contains undeclare wheat flour, an allergen.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ========================
_______________
PRODUCT:       Cecelia brand Quick Blanched Mushrooms, Pieces and Stems in 62
               ounce white plastic containers.  Recall #F-677-6.
CODE:          'JAN 09 1996' on outside label.
MANUFACTURER:  Prima Foods, Inc., Rehrersburg, New Jersey.
RECALLED BY:   Destefano Foods, Inc., Edison, New Jersey by telephone on or
               about October 16, 1995 with follow-up letter.  Firm-initiated
               recall complete.
DISTRIBUTION:  New Jersey.
QUANTITY:      93 cases were distributed.
REASON:        The mushrooms are unfit for food due to off taste and off
               odor.
                                   -2-
_______________
PRODUCT:       Preferred Selections brand Cheese Salsa in 10.75 ounce jars. 
               Recall #F-678-6.
CODE:          6 BG 103.
MANUFACTURER:  Sauces Unlimited, Inc., San Antonio, Texas.
RECALLED BY:   Manufacturer by telephone June 6, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION:  Pennsylvania, Florida, Wisconsin.
QUANTITY:      339 cases were distributed.
REASON:        The product is unfit for food due to lid seal failure and
               spoilage.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II =========================
_______________
PRODUCT:       Delflex brand and custom manufactured NMC (National Medical 
               Care) Rx peritoneal dialysis solution for home dialysis use.
               53 lots packaged in 2, 3 and 5 liter bags,
               Recall #D-231/252-
CODE:          
Fresenius product code    Product                                 Lot #
    044-50501             1.5% Dextrose, Low Mg, 5 Liter        07-008-6F 
                                                                07-162-6F
                                                                07-175-6F
                                                                07-048-6G
                                                                07-100-6G

    044-50502             2.5% Dextrose, Low Mg, 5 Liter        07-094-6F   
                                                                07-148-6F
                                                                07-183-6F
                                                                07-203-6F
                                                                07-042-6G

    044-50511             1.5% Dextrose, Std Mg, 5 Liter        07-174-6F
                                                                07-099-6G  

    044-20201             1.5% Dextrose, Low Mg, 2 Liter        07-190-6F

    044-20322             2.5% Dextrose, Low Mg, Low Ca, 2L/3L  07-193-6F

    044-50512             2.5% Dextrose, Std Mg, 5 Liter        07-196-6F
                                                                07-055-6G
                                                                07-098-6G
                                                                07-121-6G

    044-20211             1.5% Dextrose, Std Mg, 2 Liter        07-204-6F
                                                                 
    044-50524             4.25% Dextrose, Low Mg,Low Ca, 5Liter 07-079-6G 

    044-50522             2.5% Dextrose, Low Mg,Low Ca, 5 Liter 07-002-6G
                                                                07-041-6G
                                                                07-062-6G   
                              -3-                               07-063-6G                                                                07-086-6G
                                                                07-178-6G

    044-50521             1.5% Dextrose, Low Mg, Low Ca, 5 Liter07-009-6G

    044-50504             4.25% Dextrose, Low Mg, 5 Liter       07-010-6G   

                                                                07-064-6G
                                                                07-069-6G

    044-30321             1.5% Dextrose, Low Mg, Low Ca, 3 Liter07-038-6G

    044-25321             1.5% Dextrose, Low Mg, Low Ca, 2.5L/3L07-052-6G

    044-20202             2.5% Dextrose, Low Mg, 2 Liter        07-074-6G

    044-25311             1.5% Dextrose, Std Mg, 2.5L/3L        07-087-6G

    044-25314             4.25% Dextrose, Std Mg, 2.5L/3L       07-089-6G 

    044-15211             1.5% Dextrose,Std Mg, 1.5L/2L         07-147-6G

    044-20221             2.5% Dextrose, Std Mg, 2 Liter        07-154-6G   

NMC Product Code:
    65-2204-9             4.25% Dextrose, Low Mg, 2 Liter       07-121-6F
    
    65-2202-3             2.5% Dextrose, Low Mg, 2 Liter        07-189-6F
    
    65-2221-3             1.5% Dextrose, Low Mg, Low Ca, 2Liter 07-197-6F
    
    65-5502-3             2.5% Dextrose, Low Mg, 5 Liter        07-017-6G
                                                                07-034-6G
    
    65-5521-3             1.5% Dextrose, Low Mg, Low Ca, 5Liter 07-018-6G
                                                                07-078-6G
    
    65-5504-9             4.25% Dextrose, Low Mg, 5 Liter       07-049-6G
    
    65-2322-9             2.5% Dextrose, Low Mg, Low Ca, 2L/3L  07-051-6G
    
    65-5501-5             1.5% Dextrose, Low Mg, 5 Liter        07-056-6G
                                                                07-120-6G
    
    65-4302-9             2.5% Dextrose, Low Mg, 3 Liter        07-075-6G
    
    65-5522-1             2.5% Dextrose, Low Mg, Low Ca, 5Liter 07-085-6G
    
    65-3302-0             2.5% Dextrose, Low Mg, 2.5L/3L        07-088-6G
    
    65-1214-9             4.25% Dextrose, Std Mg, 1.5L/2L       07-162-6G
    
    65-3314-5             4.25% Dextrose, Std Mg, 2.5L/3L       07-163-6G
                                  -4-
MANUFACTURER:  Fresenius USA, Inc., Ogden, UT 84404
RECALLED BY:   Manufacturer by telephone August 15, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION:  Nationwide and Canada.
QUANTITY:      672,000 liters were distributed.
REASON:        Excessive endotoxin levels.

_______________
PRODUCT:       Empire Airgas labeled as "Oxygen Compressed USP" packaged in
               aluminum and steel high pressure cylinder sizes 300, 200, 125,
               E and D, Recall #D-253-6.
CODE:          All lots with 001 thru 214.
MANUFACTURER:  Empire Airgas, Inc., Syracuse, New York.
RECALLED BY:   Empire Airgas, Inc., Elmira, New York by letter August 16,
               1996.  Firm-initiated recall ongoing.
DISTRIBUTION:  New York.
QUANTITY:      31,000 units were distributed.
REASON:        GMP deficiencies.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =====================

_______________
PRODUCT:       (a) Red Blood Cells; (b) Cryoprecipitated AHF; (c) Recovered
               Plasma, Recall #B-593/595-6.
CODE:          Unit number (a-c) 30H81550.
MANUFACTURER:  American Red Cross Blood Services, Ashley, Pennsylvania.
RECALLED BY:   Manufacturer by letter December 6, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION:  Pennsylvania, New York.
QUANTITY:      1 unit was distributed.
REASON:        Blood products, which tested negative for the antibody to the
               human immunodeficiency virus type 1 (anti-HIV-1), but were
               collected from a donor who previously tested repeatedly
               reactive for anti-HIV-1, were distributed.

_______________
PRODUCT:       (a) Red Blood Cells; (b) Platelets, Recall #B-596/597-6.
CODE:          Unit numbers: (a) LH37443, KE12275, (b) KE1275.
MANUFACTURER:  Central Blood Bank, Pittsburgh, Pennsylvania.
RECALLED BY:   Manufacturer by telephone May 22 and May 31, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION:  Pennsylvania
QUANTITY       3 units were distributed.
REASON:        Blood products, collected from donors who traveled to areas
               considered endemic for malaria, were distributed.

_______________
PRODUCT:       (a) Red Blood Cells; (b) Recovered Plasma,
               Recall #B-598/599-6.
CODE:          Unit number FF37663.
MANUFACTURER:  Blood Bank of the Redwoods, Santa Rosa, California.
                                    -5-RECALLED BY:   Manufacturer by telephone and letter September 9, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION:  California.
QUANTITY:      2 unit was distributed.
REASON:        Blood products, which tested negative for antibody to the
               hepatitis C virus encoded antigen (anti-HCV), but were
               collected from a donor who previously tested repeatedly
               reactive for anti-HBC, were distributed.

_______________
PRODUCT:       Recovered Plasma, Recall #B-600-6.
CODE:          Unit number J32232.
MANUFACTURER:  San Diego Blood Bank, San Diego, California.
RECALLED BY:   Manufacturer by fax July 31, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION:  North Carolina.
QUANTITY:      1 unit was distributed.
REASON:        Blood product, collected from a donor who reported a history
               of body piercing, was distributed.

_______________
PRODUCT:       Red Blood Cells.  Recall #B-602-6.
CODE:          Unit numbers 9309726, 9309818.
MANUFACTURER:  LifeSource Blood Services, Glenview, Illinois.
RECALLED BY:   Manufacturer by telephone December 18, 1995.  Firm-initiated
               recall complete
DISTRIBUTION:  Illinois.
QUANTITY:      2 units were distributed.
REASON:        Blood products, which were exposed to unacceptable storage
               temperatures, were distributed.

_______________
PRODUCT:       (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma,
               Recall #B-603/605-6.
CODE:          Unit number (a-c) 3D2745.
MANUFACTURER:  Puget Sound Blood Center, Seattle, Washington.
RECALLED BY:   Manufacturer by letter July 13, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION:  Washington, California.
QUANTITY:      3 units were distributed.
REASON:        Blood products, collected from a donor whose spouse was
               diagnosed with hepatitis B, were distributed.

_______________
PRODUCT:       (a) Cryoprecipitated AHF; (b) Recovered Plasma, Recall #B-
               610/611-6.
CODE:          Unit number: (a & b) 38FN90184.
MANUFACTURER:  American Red Cross, Fort Wayne, Indiana.
RECALLED BY:   Manufacturer by letters March 27 and 29, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION:  Wisconsin, California.
                                    -6-QUANTITY:      2 units were distributed.
REASON:        Blood products, corresponding to a unit of Red Blood Cells
               which were contaminated with a gram positive cocci suspected
               to be Serratia Marcescens, were distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ====================

_______________
PRODUCT:       Recovered Plasma, Recall #B-586-6.
CODE:          Unit number 40GR43997.
MANUFACTURER:  American Red Cross Blood Services, Peoria, Illinois.
RECALLED BY:   Manufacturer by questionable plasma inquiry dated April 18,
               1996.  Firm-initiated recall complete.
DISTRIBUTION:  California.
QUANTITY:      1 unit was distributed.
REASON:        Blood product, which tested initially reactive for the
               hepatitis B surface antigen (HBsAg), and was not retested as
               required, was distributed.

_______________
PRODUCT:       Irradiated Red Blood Cells, Recall #B-601-6.
CODE:          Unit numbers 30FG09793, 30FH18877.
MANUFACTURER:  American Red Cross Blood Services, Ashley, Pennsylvania.
RECALLED BY:   Manufacturer by telephone on February 13, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION:  Pennsylvania
QUANTITY:      2 units were distributed.
REASON:        Blood products, labeled with an extended expiration date, were
               distributed.

_______________
PRODUCT:       Red Blood Cells, Recall #B-609-6.
CODE:          Unit numbers 013L72772, 013T72528.
MANUFACTURER:  American Red Cross, Detroit, Michigan
RECALLED BY:   Manufacturer by letter July 19, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION:  Michigan.
QUANTITY:      2 units were distributed.
REASON:        Blood products, mislabeled with the incorrect expiration date
               were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II ======================
_______________
PRODUCT:       Becton Dickinson Control Syringe, a device that consists of a
               10cc calibrated hollow plastic barrel and a movable plunger,
               Recall #Z-983-6
CODE:          Catalog number 304134, Lot #6C428, 6C429.
MANUFACTURER:  Becton Dickinson Division, North Canaan, Connecticut.
                                  -7-RECALLED BY:   Becton Dickinson, Franklin Lakes, New Jersey by telephone
               April 18, 1996, with follow-up letter of April 23, 1996. 
               Firm-intiated recall complete.
DISTRIBUTION:  Connecticut
QUANTITY:      40,900 units were distributed.
REASON:        A molding problem could result in cracked thumbring.

_______________
PRODUCT:       Biological Microroughins Latex Examination Gloves, Powder-Free
               in all sizes, Recall #Z-1054-6.
CODE:          "CM 1054 M 9403".
MANUFACTURER:  Tekmedic (M) SDN. BHD., Taman Tun Drismai, Mali.
RECALLED BY:   Intax Inc. (d/b/a Calderwood Rubber & Plastic), Santa Barbara,
               California by letter and telephone August 4, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION:  California.
QUANTITY:      655 cases were distributed.
REASON:        The gloves failed the water leak test.

_______________
PRODUCT:       Haemotronic Arterial Venous Blood Tubing Set for Hemodialysis
               Model 5M 4208, Recall #Z-1096-6.
CODE:          Lot #B942538.
MANUFACTURER:  Haemotronic S.p.A., Modena, Italy.
RECALLED BY:   Haemotronic Inc., Fairfield, New Jersey by fax on April 3,
               1996. Firm-initiated recall ongoing.
DISTRIBUTION:  Tennessee.
QUANTITY:      9,984 sets were distributed.
REASON:        The device is subject to blood leakage.

_______________
PRODUCT:       ICON Software Versions Less Than 6.0.2 used on the ICON A, AP,
               P, V, and MacICON Workstations and the Integrated Gamma Camera
               Systems Diacam, Orbiter, Multispect 2 and Multispect 3, Recall
               #Z-1097/1100-6.
CODE:          None.
MANUFACTURER:  Siemens Medical Systems, Inc., Hoffman Estates, Illinois.
RECALLED BY:   Manufacturer by letters February 1996, March 1996 and June
               1996.  Firm-initiated recall ongoing.
DISTRIBUTION:  Nationwide and international.
QUANTITY:      2,240 units were distributed.
REASON:        The ICON software versions less than 6.0.2 contained bugs
               which effect the acquisition of patient data, caused the loss
               of patient data, and require the re-injection of patient with
               radioactive isotopes in some instances.

                                 -8-
_______________
PRODUCT:       SIRECUST 960/961/1260/1261/1280/1281 MULTI I/O; Pressure
               Waveform Triggering of External Devices, Software Versions up
               to and including VA4, Recall #Z-1103/1107-6
CODE:          Part number               Description
               28 74 506 E2501           S/W UPG VA2-XMS MIOP 9+12
               28 74 514 E2501           S/W UPG VA2-AXX Multi I/O 9+12
               33 60 760 E2501           S/W UPG VBO-BXX Multi I/O 9+12
               43 27 008 E2501           S/W UPG VA4-XXX MIOP 12/9
               47 13 637 E2501           S/W UPG VBO/VA4 Multi I/O 9+12
MANUFACTURER:  Siemens Medical Systems, Inc., Danvers, Massachusetts.
RECALLED BY:   Manufacturer issued a device safety alert dated May 6, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION:  Nationwide, international.
QUANTITY:      Undetermined amount distributed.
REASON:        Misbranded.  The user manual does not address the possible
               potential for the loss trigger and for the balloon pump to
               remain in a deflated state which will produce an audible alarm
               when it occurs. 


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ======================
_______________
PRODUCT:       Linvatec brand, SPIKED WASHER, Catalog Number 8631S, 17mm X
               2.5, Sterile, Single used in conjunction with Linvatec soft
               tissue anchoring screws to attach ligaments, tendons and other
               soft tissues to long bone, Recall #Z-1095-6
CODE:          Lot Number 33747
MANUFACTURER:  Linvatec, Largo, FL 34643
RECALLED BY:   Manufacturer, by letter July 12, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION:  Nationwide.
QUANTITY:      26 devices were distributed.
REASON:        Both the package and etching on the washer was labeled as 17mm
               x 2.5 instead of the correct dimension of 17mm x 1.3.

______________


END OF ENFORCEMENT REPORT FOR September 4, 1996.  BLANK PAGES MAY FOLLOW.

                                    -9-


                                   ####