RECALLS AND FIELD CORRECTIONS

August 7, 1996                                    96-32

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Star Bake Shop Bread: (a) Italian Bread 16
               ounces; (b) Garlic Bread with Romano Cheese, 9
               ounces.  Recall #F-637/638-6.
CODE           Product manufactured through June 13, 1996.
MANUFACTURER   Leon's Bakery, North Haven, Connecticut (raw
               frozen dough).
RECALLED BY    Star Markets Company, Inc., Cambridge,
               Massachusetts, by telephone on June 13, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Massachusetts, New Hampshire.
QUANTITY       Firm estimates none remains on the market.
REASON         Products contained undeclared dry egg whites.

_______________
PRODUCT        Wavy Lays Original Potato Chips, in 6 ounce
               bags.  Recall #F-639-6.
CODE           Jul 02 96.
MANUFACTURER   Frito-Lay, Inc., Binghamton, New York.
RECALLED BY    Frito-Lay, Inc., Plano, Texas, by electronic
               mail on May 29, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Virginia, District of Columbia, New York,
               Pennsylvania, Delaware, Maryland, West
               Virginia,  New Jersey.
QUANTITY       156 cases (12 bags per case) were distributed;
               firm estimates none remains on the market.
REASON         Product contained undeclared egg yolk solids. 
               The bags actually contained Wavy Lays Hidden
               Valley Ranch flavored potato chips.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
_______________
PRODUCT        Deux Mondes brand Les Olives Candy in 3 ounce
               retail bags and Deux Mondes brand Pebbles from
               the Mountain Candy, in 3 ounce retail bags. 
               Recall #F-627/628-6.
CODE           None.
MANUFACTURER   F. Marion Chocolatier, Oraison, France.
RECALLED BY    Deux Mondes Import/Export Company,
               Minneapolis, Minnesota, by telephone on June
               28, 1996.  Firm-initiated recall complete.
DISTRIBUTION   Minnesota, Washington state.
QUANTITY       72 retail packages of each product were
               distributed.
REASON         Product contains undeclared FD&C Yellow No. 5
               and carbon black.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
_______________
PRODUCT        Blueberry Fruit Syrup, in 750 ml bottles under
               the following labels:  Naturola, Campina,
               Tropico.  Recall #F-625-6.
CODE           None.
MANUFACTURER   Caribe Tropical Corporation, San Sebastian,
               Puerto Rico.
RECALLED BY    AC Packers Corporation, Guaynabo, Puerto Rico,
               by visit July 1-5, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   Puerto Rico.
QUANTITY       180 bottles were distributed.
REASON         Product contains undeclared FD&C Blue No. 1.

_______________
PRODUCT        Minute Maid Citrus Punch, in 64 fluid ounce
               table top cartons.  Recall #F-626-6.
CODE           AUG 12 W-3L-1 ZZ:ZZ, AUG 12 W-4L-1 ZZ:ZZ
               AUG 12 W-3L-2 ZZ:ZZ, AUG 12 W-4L-2 ZZ:ZZ
               [Note - "ZZ:ZZ" refers to a 24 hour code]. 
               Case Code - 08/12/96.
MANUFACTURER   West Lynn Creamery, Lynn, Massachusetts.
RECALLED BY    Coca-Cola Foods, Houston, Texas, by visit on
               or about June 28, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Massachusetts.

                             -2-QUANTITY       240 cases (6 cartons per case) were
               distributed.
REASON         Product contains undeclared color additives
               and bears the incorrect juice content.

_______________
PRODUCT        Enchant brand Sweetened Flavorings in 3.5
               ounce canisters with tin-plate crimp bottom
               seal and plastic shaker top:
               (a) Strawberry Flavoring; (b) Hazelnut
               Flavoring; (c) Cappuccino Flavoring; (d) Peach
               Flavoring; (e) Vanilla Almond; (f) Chocolate
               Mint.  Recall #F-630/635-6.
CODE           Lot numbers:  (a) 664041; (b) 462221
               (c) 264011; (d) 862221; (e) lot 362213
               (f) 162221.
MANUFACTURER   CF Technologies, Inc., Pentwater, Michigan.
RECALLED BY    Manufacturer, by letter dated July 10, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Michigan and Ohio.
QUANTITY       Approximately 9,600 canisters were
               distributed.
REASON         Product was contained in containers with a
               defective bottom seal.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Xactdose Phenytoin Oral Suspension, USP, 100
               mg/4 ml unit dose cups, anticonvulsant. 
               Recall #D-217-6.
CODE           Lot numbers:  508608 and 508613 EXP 2/97.
MANUFACTURER   Parke-Davis, Division of Warner-Lambert
               Company, Morris Plains, New Jersey.
RECALLED BY    Xactdose, Inc., South Beloit, Illinois
               (repacker), by letter dated July 16, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,947 cases were distributed; firm estimated
               that 10-15% of the product remained on the
               market at time of recall initiation.
REASON         Due to large particle size, some of the unit
               doses may not meet potency specifications.

_______________
PRODUCT        Kanamycin Sulfate Injection USP, Rx
               aminoglycoside antibiotic for I.V. or
               intramuscular injection for the short-term
               treatment of serious infections caused by E.
               coli, Proteus sp., Enterobacter aerogenes,
               Klebsiella pneumoniae, Serratia marcescens and
               Acinetobacter sp., 10 vials per shelf carton:

                             -3-               (a) 1 g/3 ml - Catalog #06305, 3 ml single
               dose vial; (b) 250 mg/ml - Catalog #06402, 
               2 ml single dose vial.  Recall #D-218/219-6.
CODE           Lot numbers: (a) 951269, 960104, 960256; (b)
               951270, 960155, 960250.
MANUFACTURER   SoloPak Laboratories Inc., Elk Grove Village,
               Illinois.
RECALLED BY    Manufacturer, by letter dated July 26, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       154,620 vials were distributed; firm estimated
               that 84% of product remained on market at time
               of recall initiation.
REASON         Inadequate microbial controls for the raw
               material.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh
               Frozen Plasma; (d) Recovered Plasma.  
               Recall #B-554/557-6.
CODE           Unit numbers:  (a) 1878693, 6049334, 7811187;
               (b) 1878693, 7811187; (c) 1878693, 7811187;
               (d) 6049334.
MANUFACTURER   LifeSource Blood Services, Glenview, Illinois.
RECALLED BY    Manufacturer, by letter dated February 23,
               1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Illinois.
QUANTITY       (a) 3 units; (b) 2 units; (c) 2 units; 
               (d) 1 unit.
REASON         Blood products collected from a donor who
               previously tested HCV reactive and was not
               deferred. 

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c)
               Recovered Plasma.  Recall #B-565/567-6.
CODE           Unit #5009350.
MANUFACTURER   North Jersey Blood Center, East Orange, New
               Jersey.
RECALLED BY    Manufacturer, by telephone on December 21,
               1994, followed by letter dated December 22,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   New Jersey.
QUANTITY       1 unit of each component.
REASON         Blood products collected from a donor with a
               history of ear piercing within the previous 12
               months. 


                             -4-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III =====
_______________
PRODUCT        Red Blood Cells.  Recall #B-568-6. 
CODE           Unit numbers:  29135-0974 and 29135-0979.
MANUFACTURER   United Blood Services, Chicago, Illinois.
RECALLED BY    Manufacturer, by telephone on May 13, 1996,
               followed by letter dated May 31, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       2 units.
REASON         Blood products remained at room temperature
               for an unacceptable length of time. 


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Cardiac SPECT Pallet Kit, Item #1758320, used
               to conduct cardiac SPECT studies utilizing
               Technetium radionuclides.  Recall #Z-989-6.
CODE           All kits shipped between 10/01/94 and 3/7/96.
MANUFACTURER   Siemens Medical Systems, Inc., Nuclear
               Medicine Group, Hoffman Estates, Illinois.
RECALLED BY    Manufacturer, by sending copies of the
               Operators Manuals with a letter of explanation
               dated March 8, 1996.  Firm-initiated field
               correction complete.
DISTRIBUTION   Michigan, Ohio, Tennessee, Minnesota, New
               Jersey, Georgia, Missouri, California.
QUANTITY       11 units were distributed and corrected by
               supplying the missing manuals.
REASON         The cardiac pallet kits did not include the
               operating instructions manual, publication
               #17-58-155, Revision 01 (March 1992).

_______________
PRODUCT        Centurion Laser Shower System, for veterinary
               use only:  pain therapy, accelerated healing
               (biostimulation).  Recall #Z-1000-6.
CODE           None.
MANUFACTURER   Centurion Systems, Inc., Edison, New Jersey.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan July 23, 1996. Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       83 units were distributed.
REASON         Non-compliance with performance standards for
               laser products in that it lacked a remote
               interlock connector, adequate emission delay
               and certain labeling information.

                             -5-_______________
PRODUCT        Dade Actin Activated Cephaloplastin Reagent
               (APAC), for use in the determination of the
               activated partial thromboplastin time and
               other coagulation procedures requiring an
               activated partial thromboplastin reagent:      
               (a) Catalog #291011;  (b) Catalog #B4218-2; 
               (c) Catalog #B4218-999. 
               Recall #Z-1001/1003-6.
 CODE          Lot numbers:  (a) APAC-622A EXP 12/18/97;
               (b) APAC-622B EXP 12/18/97;
               (c) APAC-622C EXP 12/18/97.
MANUFACTURER   Dade International of Puerto Rico, Aguada,
               Puerto Rico.
RECALLED BY    Dade International, Inc., Miami, Florida, by
               letter January 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,281 packages were distributed.
REASON         The lots may give prolonged activated partial
               thromboplastin times with heparinized
               controls, resulting in test results higher
               than expected.

_______________
PRODUCT        Graham Field White Latex Finger Cots, sizes
               small, large and ex-large, also distributed
               under DMI label:
               (a) Graham Field White Latex Finger Cots,
               Catalog Number 88-3910S-D;
               (b) Graham Field White Latex Finger Cots,
               Catalog Number 88-3810L;
               (c) Graham Field White Latex Finger Cots,
               Catalog Number 88-3910XL;
               (d) Graham Field White Latex Finger Cots,
               Catalog Number 88-3910XL-D.  
               Recall #Z-1006/1009-6.
CODE           Catalog numbers: 88-3910L and 88-3910XL, and
               Private label numbers 88-3910S-D and
               88-3910XL-D.
MANUFACTURER   Seohung Industrial Co.,Ltd., Seoul, Korea.
RECALLED BY    Graham Field Inc., Hauppauge, New York, by
               letter dated June 26, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       5,577 gross were distributed.
REASON         The finger cots failed the FDA's WEAC analysis
               of sample #1999024 for water leaks upon entry
               into the United States.  The finger cots
               contain embedded particles, adhering surfaces,
               and fail to meet the ASTM standards for
               dimensional requirements.

                             -6-_______________
PRODUCT        Swivel Elbow without Suction Port and Ultraset
               Flex Tube Adapter Assembly, used between
               tracheal tubes and ventilator circuits to
               provide sectioning access or easy
               repositioning of the circuitry:
               (a) Swivel Elbow without Suction Port,
               Reorder/Part #DHD 66-1995, 
               (b) Ultraset Flex Tube Adapter Assembly (with
               Double Swivel Elbow without Port) Reorder/Part
               #DHD 66-2509.  Recall #Z-1010/1011-6.
CODE           Lot numbers:  (a) 613501; (b) 613108.
 MANUFACTURER  DHD Diemolding, Wampsville, New York.
RECALLED BY    DHD Diemolding Healthcare Division, Canastota,
               New York, by telephone beginning June 17,
               1996, followed by letter.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Massachusetts, New Jersey, New York,
               Pennsylvania, Minnesota, Illinois, Texas,
               California, Washington state.
QUANTITY       77 cases (50 units per case) were distributed.
REASON         The devices may contain swivel elbows with an
               open port (completely open).  

_______________
PRODUCT        Linescan System Eight A Cabinet X-ray System,
               used for security screening of baggage and
               parcels.  Recall #Z-1012-6.
CODE           Model No. SYS 108A (also known as Model No.
               LS8A-GY).
MANUFACTURER   EG&G, Astrophysics Research Corporation, Long
               Beach, California.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan May 31, 1996.  Firm-
               initiated field correction complete.
DISTRIBUTION   Nationwide.
QUANTITY       8 units.
REASON         The cabinet x-ray system failed to comply with
               the Federal cabinet x-ray system performance
               standard, in that the product did not have
               adequate indicator lights and failed to have a
               means to ensure operator presence, required
               for systems used in public areas. 


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        Coatest brand Heparin Test Kit, used for the
               determination of heparin in plasma.  
               Recall #Z-1005-6.
CODE           Kit Lot #X2513 EXP 01 May 1998, Buffer Lot
               #X1911 05 98.

                             -7-MANUFACTURER   Pharmacia Hepar, Inc., Franklin, Ohio.
RECALLED BY    Manufacturer, by letter July 19, 1996. Firm-
               initiated recall ongoing.
DISTRIBUTION   California, Florida, Wisconsin, North
               Carolina, Pennsylvania, Maine, Maryland,
               Illinois, Minnesota.
QUANTITY       39 kits were distributed.
REASON         The buffer lot number was stated incorrectly
               as X1119 on the kit carton label.

                             -8-

END OF ENFORCEMENT REPORT FOR AUGUST 7, 1996.  BLANK PAGES MAY FOLLOW.

                                   ####