FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contain s information on actions taken in connection with agency regulatory activities.
 ENFORCE
07/23/1996

 ENFORCEMENT REPORT FOR 07/24/96 

July 24, 1996                                  96-30

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Pacific Supreme Sliced Smoked Salmon, 6
               ounces.  Recall #F-621-6.
CODE           Pull Date 27 JUN 96.
MANUFACTURER   Three Star Smoked Fish Company, Los Angeles,
               California.
RECALLED BY    Manufacturer, by telephone on June 13, 1996,
               and by press release on June 21, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Arizona, California, Nevada, Oregon,
               Washington state.
QUANTITY       3,587 packages were distributed.
REASON         Product is contaminated with Listeria
               monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
_______________
PRODUCT        Mayfair Orange Slices (candy), in 30 ounce
               plastic containers.  Recall #F-612-6.
CODE           All codes. 
MANUFACTURER   Tzetzo Brothers, Inc., Buffalo, New York.
RECALLED BY    Manufacturer, by letter dated May 22, 1996,
               followed by telephone and visit.  Firm-
               initiated recall complete.
DISTRIBUTION   New York, New Jersey, Pennsylvania.
QUANTITY       109 cases were distributed. 
REASON         Product contains undeclared FD&C Yellow No. 5.

_______________
PRODUCT        Magic Farms brand Alfalfa Sprouts.  
               Recall #F-620-6.
CODE           None.
MANUFACTURER   Cool Water Farms, also known as Magic Farms,
               Salinas, California.
RECALLED BY    Manufacturer, by telephone and by letter on
               June 30, 1996.  Ongoing recall resulted from
               sample analysis and follow-up by The State of
               California FDA Branch.
DISTRIBUTION   Nevada, California.
QUANTITY       Approximately 640 pounds are involved in
               interstate commerce.
REASON         Product is contaminated with Salmonella
               montevideo.


RECALLS AND FIELD CORRECTIONS:  COSMETICS -- CLASS III ======
_______________
PRODUCT        Shampoo and Baby Bath in 8.1 fluid ounce
               bottles.  Recall #F-615-6
CODE           Catalog #30160, Item Code NA127A, Lot #BL5B1.
MANUFACTURER   Bocchi Laboratories, Inc., Walnut, California.
RECALLED BY    Shaklee U.S., Inc., San Francisco, California,
               by letter on or about July 5, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       7,635 bottles.
REASON         The shampoo and baby bath is labeled as creamy
               baby oil.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ==========
_______________
PRODUCT        Various liquid oral drug products:
                     Product                 Sizes   
               1.  Codal DH Syrup (a)       16 fluid ounces  
               2.  Cyndal HD Syrup (a)      16 fluid ounces  
               3.  Cytra-2 (a)              16 fluid ounces  
               4.  Cytra-3 (a)              16 fluid ounces 
               5.  Cytra-K (a)              16 fluid ounces 
               6.  Cytuss-HC (a)            16 fluid ounces 
               7.  Dinex Grape (a)          16 fluid ounces 
               8.  Dy-G (a)                 16 fluid ounces 
               9.  Dyline GG (b)            16 fluid ounces 
                             -2-               10. Dylix Elixir (c)         16 fluid ounces  
               11. G-Tuss (b)               1 fluid ounce   
                                            16 fluid ounces
               12. Gani Tuss DM NR (a)      16 fluid ounces
               13. Gani-Tuss NR (a)         16 fluid ounces
               14. Hist HC (b)              16 fluid ounces
               15. Hist Plus (b)            16 fluid ounces
               16. Histex (d)               16 fluid ounces
               17. Levall (d)               1 fluid ounce
                                            16 fluid ounces
               18. PC-Hist (a)              16 fluid ounces 
               19. Polytussin-XP (e)        16 fluid ounces 
               20. Tuss-DS (b)              1 fluid ounce  
                                            16 fluid ounces
               21. Tuss-HC (b)              1 fluid ounce
                                            16 fluid ounces
               22. Tuss-PD (b)              1 fluid ounce
               23. Vetuss-HC (a)            16 fluid ounces
               In addition, the following repacked capsules
               and tablets are subject to recall:
               24. Ceta-Plus Capsules (b)   25 capsules and 
                                            100 capsules 
               25. Lobac Capsules (b)       25 capsules  
                                            100 capsules  
               26. Meni-D Capsules (b)      25 capsules  
                                            100 capsules  
               27. ND Clear Capsules (b)    25 capsules  
               28. Tenake Capsules (b)      25 capsules  
               29. Versacaps Capsules (b)   25 capsules 
               30. Dyline-GG Tablets (b)    100 tablets 
               31. Gua-SR Tablets (b)       25 tablets  
                                            100 tablets
               32. Panlor Tablets (f)       100 tablets 
               33. Pannaz Tablets (f)       100 tablets  
               34. V-Dec-M Tablets (b)      25 tablets.
               Recall #D-174/207-6.
               (a)  Labeled as Marketed by: Cypress
               Pharmaceutical, Inc., Ridgeland, Mississippi.
               (b)  Seatrace Pharmaceutical brands.
               (c)  Labeled as Distributed by:  Lunsco, Inc.,
               Pulaski, Virginia 
               (d)  Labeled as Marketed by:  Salix
               Pharmaceuticals, Inc., Ridgeland, Mississippi 
               (e)  Labeled as Distributed by: Poly
               Pharmaceuticals, Inc. Mobile, Alabama 
               (f)  Labeled as Manufactured For Pan American
               Laboratories, Inc., Covington, Louisiana. 
CODE           Products 1 thru 7 -- All Lots
               8.  Lot #95001
               9. Lot numbers:  94001, 94002, 94003
               Products 10 thru 19 -- All Lots

                             -3-               20.  All Lots (1 fluid ounces), 
               Lot numbers: 95018TDS, 96049TDS (16 fluid
               ounces)
               21. All Lots (1 fluid ounces) 
               Lot numbers: 95012THC, 95020THC (16 fluid
               ounces)
               Products 22 and 23 -- All Lots
               24.  All lots (25 capsules)
               Lot #940685F (100 capsules)
               Products 25 and 26 --  All Lots
               26.  All Lots (25 capsules)
               Lot #95007MD (100 capsules)
               Products 27 thru 29 -- All Lots
               30.  Lot #96033DGT
               31.  All Lots
               32.  Lot numbers:  930821J, 95009PL
               33.  Lot #8362
               34.  All Lots.
MANUFACTURER   Seatrace Pharmaceuticals, Inc., Rainbow City,
               Alabama.
RECALLED BY    Manufacturer, by letter on July 8, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Russia.
QUANTITY       Undetermined.
REASON         Current good manufacturing practice
               deficiencies.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Ecotrin 325 mg Enteric Coated Tablets, in 4
               count blister sample packs --  non-
               prescription single ingredient, enteric coated
               aspirin.  Recall #D-208-6.
CODE           Product code: 60LAC05 (7 Lot numbers).
               Lot# 5C06A, EXP 6/96, Lot# 5D24A, EXP 7/96,
               Lot# 5G17A, EXP 9/96, Lot# 5H28A, EXP 1/97,
               Lot# 5L04A, EXP 4/97, Lot# 5I04A, EXP 1/97 
               Lot# 5L11A, EXP 4/97.
MANUFACTURING  Smith Kline Beecham/Consumer Healthcare,
               Cidra, Puerto Rico.
RECALLED BY    Smith Kline Beecham/Consumer Healthcare,
               Parsippany, New Jersey, by telephone on
               February 27, 1996, followed by letter.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       4,097 "repack" cases (72 cartons per case)
               were distributed.
REASON         Product does not meet stability
               specifications.


                             -4-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma. 
               Recall #B-512/513-6.
CODE           Unit #E07337.
MANUFACTURER   Central California Blood Center, Fresno,
               California.
RECALLED BY    Manufacturer, by letter on February 21, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from a donor who
               received a tissue transplant (surgery) within
               a year of donation, was distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Synchron CX7 Delta and Synchron CX 4CE/CX7,
               Part #469800, automated and computer
               controlled instruments designed for in-vitro
               diagnostic quantitation of biological fluid
               components and therapeutic drugs as well as
               the qualitative determination of drugs of
               abuse in urine.  Recall #Z-962-6.
CODE           All 4.0 Synchron CX Quantum IV version 4.0
               Software.
MANUFACTURER   Beckman Instruments, Brea, California.
RECALLED BY    Manufacturer, by letter on February 12, 1996. 
               Firm-initiated recall field correction
               ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       534 units were distributed.
REASON         The OFF-Line edit feature of the software is
               faulty, causing patient results to be appended
               to the record of another patient, which may
               result in inappropriate treatment.

_______________
PRODUCT        Saturne Linear Accelerators, used for the
               treatment of superficial, shallow and deep
               seated tumors.  Recall #Z-963-6.
CODE           All models except Model 800 units.
MANUFACTURER   GE Medical Systems, S.A., Buc Cedex, France.
RECALLED BY    General Electric Company, Waukesha, Wisconsin,
               by issuing field modification instructions on
               or about March 15, 1996.   Firm-initiated
               field correction complete.
DISTRIBUTION   Nationwide.
QUANTITY       36 units.
REASON         The supplemental add-on beam trimmer locking
               mechanisms can break, allowing the beam
               trimmers to fall.
                             -5-_______________
PRODUCT        Modified F-10 Culture Medium, Formula 
               #94-0155PG, 100 mL, for use in sperm
               washing/handling.  Recall #Z-969-6.
CODE           Lot #16P1349 EXP 11/30/95.
MANUFACTURER   Life Technologies, Inc., Grand Island, New
               York.
RECALLED BY    Manufacturer, by letter dated April 11, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Massachusetts.
QUANTITY       180 units were distributed.
REASON         The device contained bacterial contamination
               and was labeled "Sterile."

_______________
PRODUCT        Apelio Series Punch/Laser Combination Metal
               Working Machines, used for punching and
               cutting metal.  Recall #Z-970-6.
CODE           None.
MANUFACTURER   Amada Company, Buena Park, California.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan July 8, 1996.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Undetermined.
REASON         Some of the laser products failed to comply
               with the Federal performance standard for
               laser products, 21 CFR 1040.10 in that the
               machines lacked the required warning logotype,
               aperture label and protective housing labels;
               a product report had not been submitted; the
               user manual lacked some required information
               and quality control procedures were not
               adequate to verify compliance with the laser
               performance standard.

_______________
PRODUCT        DioLase Low Power Biostimulation Laser Model
               100, used for biostimulation, pain treatment. 
               Recall #Z-971-6.
CODE           Model 100.
MANUFACTURER   DioLase Corporation, Berkeley, California.
RECALLED BY    Manufacturer.  FDA approved the firm's
               corrective action plan July 15, 1996.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       34 units.
REASON         The device failed to comply with the Federal
               laser product performance standard, 21 CFR
               1040.10 and 1040.11, in that it lacked a
               remote interlock connector, emission delay,
               calibration procedures and schedule, labels,
               and required information.
                             -6-_______________
UPDATE         Allergan AMO Phacoflex II SI 40 NB Intraocular
               Lenses, Recall #Z-945-6, which appeared in the
               July 10, 1996 Enforcement Report should read 
               CODE:  Serial numbers: 9602010345 through
               9603004406, where the fifth digit is a '0'
               (zero)".


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        AMO Phacoflex II Intraocular Lens, 
               Model SI-30NB.  Recall #Z-964-6.
CODE           Serial numbers: 9507713263 through 9507728420,
               9508700001 through 9508744671, 9509700001
               through 9509746249, 9510700001 through
               9510751212.
MANUFACTURER   AMO-Puerto Rico, Anisco, Puerto Rico.
RECALLED BY    Allergan, Inc., Irvine, California, by letter
               on July 19, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       154,533 units were distributed.
REASON         The lenses are becoming discolored and appear
               yellow in appearance.  

_______________
PRODUCT        F-12 (HAM) Nutrient Mixture, Catalog #11765-
               047,in 1000 mL bottles, for in-vitro
               diagnostic use.  Recall #Z-965-6.
CODE           Lot #11Q4051 EXP 1/30/96.
MANUFACTURER   Life Technologies/GIBCO, Paisley, Scotland.
RECALLED BY    Life Technologies, Inc., Grand Island, New
               York, by letter dated April 28, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Michigan, Massachusetts, New Jersey,
               Pennsylvania, Alabama, California, North
               Carolina, Illinois.
QUANTITY       367 units were distributed.
REASON         Bacterial contamination.

_______________
PRODUCT        McCoy's 5A Medium Cell Culture Media, Catalog
               #16600-082, for in-vitro diagnostic use.   
               Recall #Z-966-6.
CODE           Lot #15F3844 EXP 11/30/95.
MANUFACTURER   Life Technologies/GIBCO, Paisley, Scotland.
RECALLED BY    Life Technologies, Inc., Grand Island, New
               York, by letter May 2, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide, Hong Kong, Japan.
QUANTITY       443 units were distributed.

                             -7-REASON         Bacterial contamination.

_______________
PRODUCT        RPMI-1640 Medium, 1000 mL, Catalog #21870-027,
               an in-vitro diagnostic for cell culture or
               further manufacturing uses.  Recall #Z-967-6.
CODE           14Q3744 EXP 11/30/96.
MANUFACTURER   Life Technologies/GIBCO, Paisley, Scotland.
RECALLED BY    Life Technologies, Inc., Grand Island, New
               York, by letter dated May 12, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Florida, Michigan, Pennsylvania, California,
               Ohio, Connecticut.
QUANTITY       362 units were distributed.
REASON         Bacterial contamination.

_______________
PRODUCT        Dulbecco's Modified Eagle Medium (D-MEM),
               Catalog #11885-050, in 500 mL bottles, an in-
               vitro for cell culture or further
               manufacturing uses.  Recall #Z-968-6.
CODE           Lot #22N1051 EXP 12/30/95.
MANUFACTURER   Life Technologies, Inc., Grand Island, New
               York.
RECALLED BY    Manufacturer, by letter dated April 28, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Arizona, California, Colorado, North Carolina.
QUANTITY       1,222 units were distributed.
REASON         Yeast contamination.
_______________
PRODUCT        Patient Guide Companionr 318 Nasal CPAP
               System, used for the treatment of obstructive
               sleep apnea:
               (a) Companionr 318 Nasal CPAP Home Care
               Assembly, Catalog No. 126800-00
               (b) Companionr 318 Nasal CPAP Home Care
               Assembly without the Hourmeter, Catalog No.
               126804-00.  Recall #Z-972/973-6.
CODE           Lot numbers:  (a) 512141D, 512142D, 512191B,
               601021D, 601111D, 601291D, 602121D,602271D,
               603041D, and 603131D; 
               (b) 512111B, 512181B, 512271B, 602051B,
               601081B, and 601231B.
MANUFACTURER   Lithographics, Overland Park, Kansas (patient
               guide).
RECALLED BY    Nellcor Puritan Bennett Corporation, Lenexa,
               Kansas, by letters dated May 23 or 24, 1996. 
               Firm-initiated field correction ongoing
DISTRIBUTION   Nationwide, Canada.
QUANTITY       2,247 manuals were distributed.
REASON         Some of the patient guides had sections
               omitted.
                             -8-________________
PRODUCT        Kodak T-Mat Diagnostic X-Ray films:
               (a) T-MAT G/RA, Exp. dates: 5/98 and 6/98
               Catalog No. 1295302, 35 x 43 cm
               Catalog No. 1928605, 35 x 43 cm
               Catalog No. 1373125, 35 x 43 cm
               Catalog No. 8218752, 35 x 43 cm
               Catalog No. 8301905, 35 x 35 cm
               Catalog No. 1517887, 30 x 35 cm
               Catalog No. 1510023, 24 x 30 cm
               Catalog No. 1517473, 24 x 24 cm
               Catalog No. 8378101, 14 x 51 in.
               Catalog No. 1746049, 14 x 36 in.
               Catalog No. 8261836, 10 x 12 in.
               Catalog No. 8404709, 10 x 12 in.
               Catalog No. 1756196, 8 x 10 in.
               Catalog No. 1464957, 5 x 7 in. (dental only)
               Catalog No. 1869080, 15 x 30 cm (dental only)
               (b)  T-MAT L/RA, Exp. dates: 5/98 and 6/98
               Catalog No. 1744788, 35 x 43 cm
               Catalog No. 1814631, 35 x 43 cm
               Catalog No. 8416133, 35 x 43 cm
               Catalog No. 1799980, 30 x 35 cm
               Catalog No. 1709039, 24 x 30 cm
               Catalog No. 8364911, 24 x 30 cm
               Catalog No. 8408858, 24 x 24 cm
               Catalog No. 8172413, 24 x 24 cm
               Catalog No. 8410789, 18 x 24 cm
               Catalog No. 8310286, 13 x 43 cm
               Catalog No. 8447054, 11 x 14 in
               Catalog No. 8480410, 11 x 14 in
               Catalog No. 8407710, 8 x 10 in
               Catalog No. 1163948, 8 x 10 in
               Catalog No. 1589852, 8 x 10 in (dental only)
               (c)  T-MAT S/RA, Exp. date: 5/98
               Catalog No. 1247063, 35 x 43 cm
               Catalog No. 8341422, 24 x 30 cm
               Catalog No. 8051419, 8 x 10 in
               Catalog No. 8999245 (Kodak ClinicSelect -
               CSG/1)
               (d)  T-MAT H/RA, Exp. date: 5/98
               Catalog No. 1827286, 35 x 43 cm
               Catalog No. 1085521, 35 x 35 cm
               Catalog No. 8456253, 18 x 43 cm
               Catalog No. 8210957, 14 x 51 in
               Catalog No. 8200826, 14 x 36 in
               Catalog No. 1525963, 10 x 12 in
               Catalog No. 1449206, 8 x 10 in
               Catalog No. 1067529, 15 x 30 cm (dental only)
               Catalog No. 1091057, 5 x 10 7/8 in (dental
               only)
               Catalog No. 1763044, 5 x 12 in (dental only)

                             -9-               Catalog No. 1975747 (Kodak ClinicSelect - 
               CSG/2)
               Catalog No. 8202988 (Kodak ClinicSelect -
               CSG/2)
               Catalog No. 8793846 (Kodak ClinicSelect -
               CSG/2).  Recall #Z-974/977-6.
CODE           Approximately 600 emulsion ID numbers are
               involved.
MANUFACTURER   Eastman Kodak Company, Windsor, Colorado.
RECALLED BY    Eastman Kodak Company, Rochester, New York, by
               E-mail dated June 7 and 20, 1996, by letters
               of June 11 and 14, 1996.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       55,729 boxes were distributed.
REASON         The firm's internal product testing revealed a
               small quantity of the subject products could
               exhibit a series of plus density artifacts in
               the form of round spots and straight lines due
               to static discharge in the manufacturing
               process.  There is a possibility that these
               artifacts could necessitate a repeat of a
               patient exposure.

                            -10-


END OF ENFORCEMENT REPORT FOR JULY 24, 1996.  BLANK PAGES MAY FOLLOW.

                                   ####