FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
03/27/1996

 ENFORCEMENT REPORT FOR 03/27/96 


March 27, 1996                                    96-13

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
_______________
PRODUCT        Bulk Toleron Food Grade (brand of Ferrous
               Fumarate USP), in 110 pound drums, for use as
               a nutrient and vitamin supplement.  
               Recall #F-345-6.
CODE           Lot #SPK334.
MANUFACTURER   Nitrokemia Company, Hungary.
RECALLED BY    Mallinckrodt Chemical, Inc., St. Louis,
               Missouri, by letters dated December 21, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   New Jersey, Michigan, California, Missouri,
               Minnesota.
QUANTITY       320 drums were distributed.
REASON         Product contains foreign material.

_______________
PRODUCT        Dietary supplement labeled in part,
               Performance Plus/KSA Inc. Tone and Shape Your
               Body All Natural Fat and Cellulite Burning
               Formula, in 120 capsule bottles.
               Recall #F-346-6.
CODE           Lot numbers:  7306, 7479, 1690.
MANUFACTURER   Magno-Humphries Laboratories, Inc., Tigard,
               Oregon.
RECALLED BY    Manufacturer, by telephone on February 20,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Washington state, Oregon.
QUANTITY       2,272 bottles were distributed.
REASON         The product's label declares 60 mg of iron
               when the product is formulated to contain 6 mg
               of iron.

_______________
PRODUCT        Minute Maid fruit juice products containing
               apple juice:  (a) Apple Juice in 8.45 fluid
               ounce tetrapak packages; (b) Grape Juice in
               8.45 fluid ounce tetrapak packages; (c)
               Variety Pak containing Apple Juice and Grape
               Juice in 8.45 fluid ounce tetrapak packages.  
               Recall #F-347/349-6.
CODE           Lot numbers:  (a) D21999A, D21999B, H12948C,
               H12948D, H13055B, H13055C, H13055D;
               (b) D21981E, H13235A, H13235B;
               (c) H13125B, H13220C, H13304B, H13359B,
               H13502E, H13502D, H13576E. 
MANUFACTURER   Coca-Cola Foods, in Aburndale, Florida and
               Hightstown, New Jersey.
RECALLED BY    Coca-Cola Foods, Houston, Texas, by visit on
               March 7, 1996.  Firm-initiated recall ongoing. 
               See also FDA talk paper T96-14, dated February
               26, 1996.
DISTRIBUTION   Nationwide. 
QUANTITY       57,230 cases (27 units per case) were
               distributed; firm estimated that less than 10
               percent remained on market at time of recall
               initiation.
REASON         Hydrolyzed inulin syrup has been substituted
               in part for apple juice in these products.

_______________
PRODUCT        Minute Maid and Minute Maid Orchard's Best
               fruit juice products containing apple juice as
               follows:
               (a) Minute Maid Apple Juice in 8.45 fluid
               ounce tetrapak packages;
               (b) Minute Maid Grape Juice in 8.45 fluid
               ounce tetrapak packages; 
               (c) Minute Maid Orchard's Best Country Apple
               Blackberry juice in 46 fluid ounce plastic
               containers;
               (d) Minute Maid Orchard's Best Country Apple
               Cranberry Raspberry juice in 12 ounce frozen
               cans; 
               (e) Minute Maid Orchard's Best Ripe Apple
               Raspberry juice in 12 ounce frozen cans;

                             -2-               (f) Minute Maid Orchard's Best Country Ripe
               Apple Raspberry juice in 64 fluid ounce paper
               cartons, chilled;
               (g) Minute Maid Orchard's Best Country
               Raspberry Peach juice in 46 fluid ounce
               plastic containers;
               (h) Minute Maid Orchard's Best Country Sweet
               Strawberry Peach juice in 12 ounce frozen
               cans;
               (i) Minute Maid Orchard's Best Country Sweet
               Strawberry Peach juice in 64 fluid ounce paper
               cartons, chilled; 
               (j) Minute Maid Orchard's Best Country Wild
               Mountain Grape in juice 12 ounce frozen cans;
               (k) Minute Maid Variety Pack containing in
               part Apple Juice in 8.45 ounce tetrapak
               packages and Grape Juice in 8.45 ounce
               tetrapak packages.  Recall #F-350/360-6. 
CODE           Lot numbers:  (a) T28036C, T28521A; (b)
               G13470A, G14084C; (c) D28376D, T27806C,
               T28376D, T29060E, T29061A  (d) D23409B; (e)
               D22254A, D22254B, D22254G, D23408A, D23408B,
               D23408C, D23409C; (f) H14610F, H14905D; (g)
               T26217A, T272930, T27810E, T27956B, T29061A;
               (h) D22256A, D22256B, D23410C, D23675B; (i)
               H15008A, H14609F, H14905D, H14917A; (j)
               D23421C; (k) Lots G13413A, G13479B, G13481A,
               G13483B, G13485D, G13560A, G13598A, G13642C,
               G13689C, G13691D, G14015C, G14019D, G14084D,
               G14118B, G14122E, G14259A, G14261E, T28009B,
               T28049E, T28256A, T28341E.
MANUFACTURER   Coca-Cola Foods, Auburndale, Florida;
               Coca-Cola Foods, Geneva, Ohio; 
               Coca-Cola Foods, Paw Paw, Michigan;
               Coca-Cola Foods, Hightstown, New Jersey. 
RECALLED BY    Coca-Cola Foods, Houston, Texas, by visit on
               February 23, 1996.  Firm-initiated recall
               ongoing.  See also FDA talk paper T96-14,
               dated February 26, 1996.
DISTRIBUTION   Nationwide.
QUANTITY       236,000 cases (27 units per case) were
               distributed; firm estimated that 122,000 cases
               remained on market at time of recall
               initiation.
REASON         Hydrolyzed inulin syrup has been substituted
               in part for apple juice in these products.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Dilantin Kapseals, 100 mg, Extended Phenytoin
               Sodium Capsules, in bottles of 100 and 1,000. 
               Recall #D-095-6.
                             -3-CODE           Lot numbers: 06365F EXP 6/9 and 02975F EXP
               7/97.
MANUFACTURER   Warner Lambert Company, Fajardo, Puerto Rico.
RECALLED BY    The Parke-Davis Division of Warner Lambert
               Company, Morris Plains, New Jersey, by letter
               on February 22, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       39,393 bottles of lot 06365F, and 3,952
               bottles of lot 02975F.
REASON         Product does not meet dissolution
               specifications.

_______________
PRODUCT        Metoclopramide Tablets USP, 5 mg, under either
               the Vangard Labs label, packaged 10 tablets
               per card, 10 cards per box, or the Glasgow
               Pharmaceutical label, packaged 25 tablets per
               card, 30 cards per box, for use with patients
               with symptomatic, documented gastroesophageal
               reflux.  Recall #D-097-6.
CODE           Lot numbers:  3456-005 EXP 3/97 (Vangard),
               121-002 EXP 3/97 (Glasgow).
MANUFACTURER   Biocraft Laboratories, Inc., Fair Lawn, New
               Jersey.
RECALLED BY    Vangard Labs, Inc., Glasgow, Kentucky, by
               letters on January 30, 1996 and March 13,
               1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       288 boxes (28,800 tablets) Vangard label and
               102 boxes (76,500 tablet) Glasgow label were
               distributed.
REASON         Product does not meet content uniformity
               specifications.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Cyproheptadine HCl 2 mg/5 ml Syrup, packaged
               in 16 ounce and in 1 gallon plastic
               containers, Rx antihistaminic/antiserotonergic
               agent, under the following labels:  Halsey,
               Qualitest, Schein, Aligen, Goldline.
               Recall #D-096-6.
CODE           Lot numbers:  45959 EXP 2/96, 45977 EXP 4/96,
               46008 EXP 7/96, 4H31 EXP 8/96, 4H31D 8/96,
               4K19B EXP 9/96, 5H08B EXP 6/97.
MANUFACTURER   Halsey Drug Company, Inc., Brooklyn, New York.
RECALLED BY    Manufacturer, by letter sent on March 12,
               1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Puerto Rico.

                             -4-QUANTITY       37,414 16-ounce containers and 4,624 gallon
               containers were distributed.
REASON         Assay and pH stability results are below
               specification.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT        Platelets, Pheresis.  Recall #B-278-6.
CODE           Unit #KZ24581.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by telephone on or about
               September 26, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Maryland.
QUANTITY       1 unit split into two portions.
REASON         Blood product, collected from a donor who had
               traveled to an area designated as endemic for
               malaria, was distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-293-6.
CODE           Unit numbers KZ28403 and KZ28418.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by telephone on or about August
               14, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Virginia.
QUANTITY       2 units.
REASON         Blood products, which were collected from
               donors whose health history screening was
               inadequately performed, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  
               Recall #B-331/332-6.
CODE           Unit numbers:  (a) 10494-5704, 10497-6601,
               10500-4510; (b) 10494-5704, 10500-4510.
MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, Scottsdale, Arizona.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated December 26, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Arizona.
QUANTITY       (a) 3 units; (b) 2 units.
REASON         Blood products, collected from a donor who
               lived in an area designated as endemic for
               malaria, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-333-6.
CODE           Unit numbers:  10496-7357 and 10499-2885.

                             -5-MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, Scottsdale, Arizona.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated December 21, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Arizona, Texas.
QUANTITY       2 units.
REASON         Blood products, collected from a donor who
               lived in an area designated as endemic for
               malaria, were distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT        Recovered Plasma.  Recall #B-294-6.
CODE           Unit #KN27273.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by fax on or about July 19,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product, collected from a donor whose
               health history screening was inadequately
               performed, was distributed.

_______________
PRODUCT        Fresh Frozen Plasma.  Recall #B-334-6.
CODE           Unit #10500-9485.
MANUFACTURER   Blood Systems, Inc., doing business as United
               Blood Services, Scottsdale, Arizona.
RECALLED BY    Blood Systems, Scottsdale, Arizona, by
               telephone on November 27, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Arizona.
QUANTITY       1 unit.
REASON         Blood product which was not frozen within
               eight hours of collection was distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Servo 300 Ventilator, Product #6079391 E338E. 
               Recall #Z-499-6.
CODE           Serial numbers up to and including 6087.
MANUFACTURER   Siemens-Elema AB, Solna, Sweden.
RECALLED BY    Siemens Medical Systems (SMS), Inc., Danvers,
               Massachusetts, by letter dated October 17,
               1995.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       240 units were distributed.

                             -6-REASON         The internal batteries of the device are
               subject to short circuits.

_______________
PRODUCT        Medi-Trace 530 Series Electrodes (also known
               as ECL 530 Medi-Trace Foam), ECG Conductive
               Adhesive Monitoring Electrodes.  
               Recall #Z-500-6.
CODE           All lots.
MANUFACTURER   Graphic Controls Corporation, Gananoque,
               Ontario, Canada.
RECALLED BY    Graphic Controls Corporation, Buffalo, New
               York, by "product advisory" letter issued on
               February 6, 1996.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       10,264,680 electrodes were distributed; firm
               estimated that 1,362,000 units remained on
               market at time of recall initiation.
REASON         The product's labeling does not provide
               adequate directions for use and does not
               include all necessary precaution statements.

_______________
PRODUCT        Medi-Trace 533 AMR Foam ECG Conductive
               Adhesive Electrodes.  Recall #Z-505-6.
CODE           All lots.
MANUFACTURER   Graphic Controls Corporation, Gananoque,
               Ontario, Canada.
RECALLED BY    Graphic Controls Corporation, Buffalo, New
               York, by letter dated February 9, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   California, Illinois, Kansas, Missouri,
               Louisiana, Oregon, South Carolina, Washington
               state.
QUANTITY       221,760 electrodes were distributed; firm
               estimated that 65,520 electrodes remained on
               market at time of recall initiation.
REASON         The product's labeling does not provide
               adequate directions for use and does not
               include all necessary precaution statements.

_______________
PRODUCT        Medi-Trace 530 Series Conductive Adhesive
               Electrodes used in telemetry kits:
               (a) Medi-Trace MPK-MOD2 (formerly MPK-MEM5),
               Reorder/Part No. 31027983;
               (b) Medi-Trace MPK-MOD3 (formerly MPK-MEMSW),
               Reorder/Part No. 31027942;
               (c) Medi-Trace MPK-MOD4, Reorder No. 31098448.
               Recall #Z-506/508-6.
CODE           All lots.

                             -7-MANUFACTURER   Graphic Controls Corporation, Gananoque,
               Ontario, Canada.
RECALLED BY    Graphic Controls Corporation, Buffalo, New
               York, by letter dated February 9, 1996.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Texas, Arkansas, Alabama.
QUANTITY       9,244 kits were distributed.
REASON         The product's labeling does not provide
               adequate directions for use and does not
               include all necessary precaution statements. 
               Also kits MPK-MOD2 (MPK-MEM5) and MPK-MOD3
               (MPK-MEMSW) may contain LT4 Wishard Electrodes
               which conflicts with product literature that
               states the kits contain Medi-Trace 530 series
               electrodes. 

_______________
PRODUCT        Electrosurgical generators:
               (a) Force FX; (b) Force 300, (c) NS2000. 
               Recall #Z-509/511-6.
CODE           Serial numbers:  (a) F5G101A - F6B657A; (b)
               F5I101B - F6A284B; (c) N5D101S - N5L178S.
MANUFACTURER   Valleylab, Inc., Boulder, Colorado.
RECALLED BY    Manufacturer, by letter on February 21, 1996. 
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       559 generators.
REASON         Software logic error.  A potential exists for
               allowing the output of the generator to remain
               active ("latch on") after the surgeon releases
               the button or footswitch to deactivate the
               generator.

_______________
PRODUCT        Force 10, 20, 30, and 40 Electrosurgical
               Generators.  Recall #Z-512/515-6.
CODE           All model numbers within the following serial
               number ranges except those which have been
               previously subjected to the calibration
               verification test per 10/93 Service Bulletin.
               R2L1559S - R2L1673S       R4A4044S - R4A4213S
               R3A1674S - R3A1843S       R4B4214S - R4B4388S
               R3B1844S - R3B2018S       R4C4389S - R4C4658S
               R3C2019S - R3C2288S       R4D4659S - R4D4883S
               R3D2289S - R3D2448S       R4E4884S - R4E5093S
               R3E2449S - R3E2608S       R4F5094S - R4F5298S
               R3F2609S - R3F2683S       R4G5299S - R4G5432S
               R3G2684S - R3G2873S       R4H5433S - R4H5554S
               R3H2874S - R3H3143S       R4I5555S - R4I5726S
               R3I3144S - R3I3463S       R4J5727S - R4J5924S
               R3J3464S - R3J3573S       R4K5925S - R4K6116S
               R3K3574S - R3K3733S       R4L6117S - R4L6230S

                             -8-               R3L3734S - R3L4043S       R5A6231S - R5A6374S
               R5B6375S - R5B6480S.        
MANUFACTURER   Valleylab, Inc., Boulder, Colorado.
RECALLED BY    Manufacturer, by letter January 1996. Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       4,921 generators.
REASON         A REM transformer component defect. A
               potential exists that may allow the REM alarm
               indicator light to remain green, indicating
               proper attachment of the return electrode,
               when incomplete patient contact actually
               exists.  When this occurs, the REM system will
               not alarm or disable the generator.

_______________
PRODUCT        Maxima Oxygenator product line of hollow fiber
               oxygenators, intended for use in procedures
               requiring the extracorporeal oxygenation of
               and carbondioxide removal from blood during
               cardiopulmonary bypass: 
               (a) Model No. 1380, Maxima;
               (b) Model No. CB1380, Carmeda Maxima;
               (c) Model No. 1300, Maxima with Filtered
               Reservoir;
               (d) Model No. 3380, Maxima Plus;
               (e) Model No. CB3380, Carmeda Maxima Plus;
               (f) Model No. 3300, Maxima Plus with Filtered
               Reservoir;
               (g) Model No. 3310, Maxima Plus with
               non-filtered Reservoir;
               (h) Model No. MAX-PRF, Maxima Plus PRF;
               (i) Model No. CBMAX-PRF, Carmeda Maxima Plus
               PRF;
               (j) Model No. MAX-PRFB, Maxima Plus PRF with
               filtered Reservoir;
               (k) Model No. MAX-PRFT, Maxima Plus PRF with
               filtered Reservoir;
               (l) Model No. MAX-PRFN, Maxima Plus PRF with
               non-filtered Reservoir;
               (m) Model No. 3311, Maxima Plus with Hardshell
               Venous  Reservoir.  Recall #Z-516/528-6.
CODE           All lot numbers. 
MANUFACTURER   United For Excellence, Inc. (UFE), Riverfalls,
               Wisconsin (Hex housing component).
RECALLED BY    Medtronic Inc., Cardiopulmonary Division,
               Anaheim, California, by sending inter-office
               memorandum on October 24, 1995, followed by
               letter January 19, 1996.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       56,125 units were distributed.

                             -9-REASON         The venous inlet port is subject to leakage
               due to defective molding.

_______________
PRODUCT        Programalith III Pulse Generators:
               (a) Model 241, unipolar with a lead connector
               size of 5 mm
               (b) Model 242, bipolar and accepts a
               bifurcated lead
               (c) Model 249, unipolar with a lead connector
               size of 6 mm.  Recall #Z-529/531-6.
CODE           Serial numbers 86900 to 87949, 107800 to
               107899, and 108000 to 108866.
MANUFACTURER   Pacesetter, Inc., Sylmar, California.
RECALLED BY    Manufacturer, by letter February 15, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Japan, Latin America.
QUANTITY       1,290 units were distributed.
REASON         Devices may fail due to output loss.

_______________
PRODUCT        Force 10, 30, and 40 Electrosurgical
               Generators.  Recall #Z-548/550-6.
CODE           All of the following serial numbers:
               R5L8276S   R5L8199S   R5L8164S   R6B8282S
               R5L8280S   R5L8200S   R5L8165S   R6B8283S
               R5J8016S   R5L8201S   R5L8168S   R6B8291S
               R5J8017S   R5L8204S   R5L8170S   R6B8293S
               R5J8018S   R5L8248S   R5L8171S   R6B8294S
               R5J8019S   R5L8253S   R5L8172S   R6B8295S
               R5J8021S   R5L8254S   R5L8173S   R6B8297S
               R5J8023S   R5L8255S   R5L8174S   R6B8298S
               R5J8024S   R5L8256S   R5L8175S   R6B8300S
               R5J8025S   R5L8257S   R5L8176S   R6B8301S
               R5K8050S   R5L8258S   R5L8177S   R6B8302S
               R5K8067S   R5L8259S   R5L8178S   R6B8343S
               R5K8068S   R6B8303S   R5L8179S   R6B8344S
               R5K8069S   R6B8304S   R5L8180S   R6B8345S
               R5K8070S   R6B8305S   R5L8184S   R6B8346S
               R5L8260S   R6B8306S   R5L8185S   R6B8350S
               R5L8261S   R6B8307S   R5L8190S   R6B8352S
               R5K8149S   R6B8308S   R5L8191S   R6B8353S
               R5K8150S   R6B8371S   R5L8192S   R6B8354S
               R5K8151S   R6B8372S   R5L8193S   R6B8355S
               R5K8152S   R5L8161S   R5L8194S   R6B8359S
               R5K8153S   R5L8162S   R5L8195S   R6B8360S
               R5L8198S   R5L8163S   R5L8196S   R6B8361S.
MANUFACTURER   Valleylab, Inc., Boulder, Colorado.
RECALLED BY    Manufacturer, by letter dated March 7, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       93 generators.

                            -10-REASON         Some of the generators contain a defective
               component that has the potential for
               initiating self-activation, which could result
               in burns to the patient or surgical personnel.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        High Definition Imaging Ultrasound System HDI
               3000 Part #8500-0030-01, projects ultrasound
               energy from a transducer in direct contact
               with the body for the purpose of detecting
               abnormalities in soft tissue structures and
               vessels.  Recall #Z-428-6.
CODE           Domestic serial numbers:
               003GC1    003GBT    003GBQ      003GGN
               003GDJ    003GG8    003GFV      003GC5
               003GC0    003GCR    003GC4      003GGK
               003GG7    003GFZ    003GG4      003GBP
               003GDH    003GBR    003GC2      003GCT 
               003GFW    003GDC         
               International serial numbers:
               003GF1    003GDM    003GBX      003GCC
               003GF5    003GBY    003GCB      003GDL 
               003GGV    003GGW    003GCD      003GCF
               003GDW    003GF9    003GGY      003GDT 
               003GCJ.         
MANUFACTURER   Advanced Technology Laboratories, Bothell,
               Washington.
RECALLED BY    Manufacturer, by letters dated August 14 and
               23, 1995.  Firm-initiated field correction
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       39 units were distributed
REASON         A software coding error causes a calculation
               error that affects  the accuracy of the color
               M-Mode time measurements.

_______________
PRODUCT        High Definition Imaging Ultrasound System HDI
               3000 Part No. 8500-0030-01 with 8.2.2 and
               8.2.3 Software Versions, projects ultrasound
               energy from a transducer in direct contact
               with the body for the purpose of detecting
               abnormalities in soft tissue structures and
               vessels.  Recall #Z-501-6.
CODE           HDI 3000 systems and 8.2.2 and 8.2.3 upgrades.
MANUFACTURER   Advanced Technology Laboratories (ATL),
               Bothell, Washington.
RECALLED BY    Manufacturer, by letter dated January 22,
               1996.  Firm-initiated field correction
               complete.

                            -11-DISTRIBUTION   Nationwide and international.
QUANTITY       344 units subject to field correction.
REASON         An error in the software which caused the
               Doppler history buffer to become discontinuous
               and out of synch with time.

                            -12-

END OF ENFORCEMENT REPORT FOR MARCH 27, 1996.  BLANK PAGES MAY
FOLLOW.

                                   ####