FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
02/21/1996

 ENFORCEMENT REPORT FOR 02/21/96 


February 21, 1996                                 96-08

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Dilantin, extended Phenytoin Sodium Capsules,
               USP), Kapseals, 100 mg, antiepileptic drug, in
               bottles of 100.  Recall #D-086-6.
CODE           Lot #09054FA EXP 5/96.
MANUFACTURER   Warner Lambert Company, Fajardo, Puerto Rico.
RECALLED BY    Parke Davis, Division of Warner Lambert,
               Morris Plain, New Jersey, by letter December
               12, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       39,846 bottles were distributed.
REASON         Product may not meet dissolution
               specifications through expiration date.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Panama Jack brand Dark Tanning Lotion,
               waterproof, 4 SPF, in 3 fluid ounce plastic
               bottles.  Recall #D-085-6.
CODE           5M01A.
MANUFACTURER   Creative Cosmetics, Rockledge, Florida.
RECALLED BY    Manufacturer, by letter April 3, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       7,344 units were distributed.
REASON         Product does not meet manufacturer's microbial
               specifications.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Multinex, a breath-by breath gas analyzer that
               uses microprocessor technology to calculate
               and display:  End-Tidal CO2, Inspired CO2,
               Inspired and Expired N2O, Inspired 02,
               Inspired and Expired Anesthetic Agents,
               Respiration Rate, Sa)2, and Pulse Rate:  (a)
               Multinex Plus, Model Number 0998-00-0071-13,
               14, 15, 16; (b) Multinex Plus w/ID, Model
               0998-00-0071-17, 18, 19, 20; (c) Multinex,
               Model Number 0998-00-0071-01, 02, 03, 04; (d)
               Multinex ID, Model 0998-00-0071-09, 10, 11,
               12.  Recall #Z-326/329-6.
CODE           All Products contain the 4610 Gas Analyzer
               Board and the 4640 Gas Identification Board.
MANUFACTURER   Datascope Corp., Paramus, New Jersey.
RECALLED BY    Manufacturer, by letter sent beginning August
               31, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,248 units were distributed.
REASON         The 4610 Gas Analyzer Board and the 4640 Gas
               Identification Board have been found to have
               defects which may cause a random and
               intermittent failure message along with either
               a temporary loss of gas monitoring or a
               disabling of the Anesthesia Identification
               function.  Units with anesthesia agent
               identification also have the potential to
               mis-identify an agent.

_______________
PRODUCT        Monogram Instrument Tibial Baseplate
               Impactor/Extractor, used to install and remove
               the tibial baseplate.  Recall #Z-341-6.
CODE           Catalog #6633-7-575.
MANUFACTURER   Howmedica, Inc., Rutherford, New Jersey.
RECALLED BY    Manufacturer, by letter dated March 14, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       408 units were distributed.
REASON         The product's threaded shaft was breaking
               during impaction.

                             -2-_______________
PRODUCT        (a) Alta Tibial Rod Driver Locking Bolt (bolt
               only), used for the insertion of the Alta
               tibial rod; (b) Alta Tibial Rod Driver and
               Bolt (driver/bolt assembly) meant to absorb
               the impact of the blows from the mallet during
               insertion of the tibial rod.  Recall #Z-
               344/345-6.
CODE           Catalog numbers:  (a) 5255-2-697; 
               (b) 5255-2-690.
MANUFACTURER   Howmedica, Inc., Rutherford, New Jersey.
RECALLED BY    Manufacturer, by letter dated July 29, 1994. 
               Firm-initiated complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       709 units were distributed; firm estimates
               none remains on the market.
REASON         The locking bolt was breaking near the
               threaded end - smooth shaft interface during
               surgical procedures.

_______________
PRODUCT        Barrier Vu-Thru Cardiovascular Pack:
               (a) Sterile BARRIER Vu-Thru Cardiovascular
               Pack containing in part a C-V Split Sheet,
               Product (Reorder) Number 127; 
               (b) Sterile BARRIER Vu-Thru Cardiovascular
               Pack II containing in part a C-V Split Sheet,
               Product (Reorder) Number 0173, (
               (c) Bulk non-sterile BARRIER Vu-Thru C-V/Split
               Sheet, Product (Reorder) Number 9573.
               Recall #Z-374/376-6.
CODE           Lot numbers:  (a) S01620 through S01713
               (b) S00170 through S00193
               (c) S00440 through S00471.
MANUFACTURER   Johnson & Johnson Medical, Inc., El Paso,
               Texas.
RECALLED BY    Johnson & Johnson Medical (J&J), Inc.,
               Arlington, Texas, by letters dated December 8,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       569 cases (2,770 units) were distributed.
REASON         The fabric of the clipping tabs are subject to
               tearing, presenting potential harm to patient.
               

_______________
PRODUCT        (a) Gesco brand 3.5 Fr. Umbili-Cath, a
               sterile, single-use, prescription PVC
               umbilical vessel catheter; (b) Umbili-Cath
               Tray, a procedural tray for the insertion of
               umbilical vessel catheters.
               Recall #Z-379/380-6.

                             -3-CODE           (a) Catalog No. UC 9823, Lot No.  06-112396;
               (b) Catalog No. UCT 9823, Lot No. 26-024296.
MANUFACTURER   Gesco International, Inc., subsidiary of
               MedChem Products, Inc., San Antonio, Texas.
RECALLED BY    MedChem Products, Inc., Woburn, Massachusetts,
               by letters dated May 19, 1995.  FDA-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 900 units; (b) 700 units were distributed;
               firm estimates none remains on market.
REASON         The catheter was manufactured without a
               radiopaque stripe, thus precluding
               radiographic confirmation. 

_______________
PRODUCT        Rite-Angle Disposable Prophy Angle with Soft
               Cup, used in performing dental prophylaxis.
               Recall #Z-408-6.
CODE           Lot numbers:  950108-1, 950110-2, 950118-1,
               950119-1, 950119-2.
MANUFACTURER   Team Technologies, Inc., Morristown,
               Tennessee.
RECALLED BY    Dentsply International, York, Pennsylvania, by
               letter April 26, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       112,320 were distributed.
REASON         In some cases the device has demonstrated a
               propensity for stalling, stopping, falling
               apart, vibrating, freezing-up, and over-
               heating during usage.

_______________
PRODUCT        Power Sonic brand Battery Packs used in Uro-
               View or UroView Model 2000 X-ray Imaging
               Systems.  Recall #Z-409-6.
CODE           All Power Sonic brand batteries received
               during 1995 and shipped to the field as
               replacements.
MANUFACTURER   OEC Medical Systems, Inc., Salt Lake City,
               Utah.
RECALLED BY    Manufacturer, by letter on December 14, 1995. 
               All of the affected batteries were replaced
               between 12/15/95 and 12/28/95.  Firm-initiated
               recall complete.
DISTRIBUTION   Florida, Illinois, Kentucky, North Carolina,
               New York, Ohio, Virginia, Washington, D.C.
QUANTITY       8 units.
REASON         The replacement battery pack could have been
               damaged during shipment which could possibly
               lead to premature failure, and in the worst
               case, a change of a battery fire.

                             -4-_______________
PRODUCT        Dextrolyte Peritoneal Dialysis 5-prong Cycler
               Set.  Recall #Z-411-6.
CODE           Catalog #48-1110-5, Lot R5A516.
MANUFACTURER   National Medical Care - Reynosa, Reynosa,
               Mexico.
RECALLED BY    National Medical Care, Rockleigh, New Jersey,
               by letter May 26, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       2,990 sets were distributed.
REASON         The connector lines are misassembled.

_______________
PRODUCT        Baxter Arterial Bloodline, used for dialysis. 
               Recall #Z-412-6.
CODE           Catalog #5M4253N, Lot #R4H604.
MANUFACTURER   National Medical Care - Reynosa, Reynosa,
               Mexico.
RECALLED BY    National Medical Care, Rockleigh, New Jersey,
               by letter June 2, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       725 cases were distributed.
REASON         The arterial bloodlines were packaged in
               cartons labeled as venous bloodlines.

_______________
PRODUCT        (a) Stryker Universal Cement Restrictor with
               Inserter, Part #206-701, intended to be
               implanted during total joint or joint revision
               arthroplasty; (b) Stryker Sterile, disposable
               Cement Sculps, Part #206-716, intended to be
               used in joint or joint revision arthroplasty
               to remove excess bone cement from around the
               prosthesis.  Recall #Z-429/430-6.
CODE           Lot numbers:  (a) 95071402S; (b) 95071382S.
MANUFACTURER   BioMedical Devices, Irvine, California.
RECALLED BY    Stryker Instruments, Division of Stryker
               Corporation, Kalamazoo, Michigan, by letter
               dated August 24, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 60 boxes (6 units per box); (b) 32 boxes
               (6 units per box) were distributed.
REASON         The sterility of the product cannot be
               guaranteed because the packaging may be
               compromised.

_______________
PRODUCT        Belmont Acuray 071A Dental Periapical X-Ray
               System.  Recall #Z-433-6.
CODE           Serial numbers:  M007051R to 007150R.
                             -5-MANUFACTURER   Takara Medical Company, Kobe, Japan.
RECALLED BY    Belmont Equipment Corporation, Somerset, New
               Jersey, by letters of September 20, 1995, and
               October 27, 1995.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       3,851 units were distributed.
REASON         The horizontal arm of the device may separate
               from its wall mount bracket and fall.  In
               addition, the horizontal arm may rise
               vertically in its wall mount bracket, but did
               not disengage.

_______________
PRODUCT        Radiographic Intensifying Screens, intended to
               expose medical radiographic film:
               (a) Model Trimax T2; (b) Model Trimax T6; (c)
               Model Trimax T12; (d) Model Trimax T4; (e)
               Model Trimax T8; (f) Model Trimax T6/12.
               Recall #Z-435/440-6.
CODE           All serial numbers.
MANUFACTURER   3M Italia SPA, Viale Della Liberta, Italy.
RECALLED BY    3M Medical Imaging Systems Division, St. Paul,
               Minnesota, by letter sent on November 17,
               1995, and by electronic mail message November
               21, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       39,150 screens were distributed.
REASON         The intensifying screens will chemically react
               with radiographic film and become heavily
               discolored.  The discolored screens will
               result in decreasing the film speed and
               increasing patient exposure to ionizing
               radiation during a radiographic patient
               exposure to ionizing radiation during a
               radiographic examination.

_______________
PRODUCT        ProSpecT(tm) Entamoeba Histolytic Microplate
               Assay Kit, used for detection of Entamoeba
               Histolytica specific antigens.
               Recall #Z-443-6.
CODE           Catalog #560-96, Lot numbers 940491 and
               950023.
MANUFACTURER   Alekon Biomedical, Inc., Sunnyvale,
               California.
RECALLED BY    Manufacturer, by fax on or about March 23,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       281 kits were distributed.
REASON         The test could provide a false positive result
               for Entamoeba histolytic in pediatric samples.

                             -6-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        ATI EO Gas Disposable Test Pack with ATI-Test
               Biological Indicator, a disposable test pack
               with a self-contained biological indicator
               used to monitor ethylene oxide sterilization
               cycles: (a) Catalog #00234/00234BX;
               (b) Catalog #23404/23404BX.
               Recall #Z-385/386-6.
CODE           Lot numbers:  (a) 9501940702, 9501940703,
               9504931002, 9504940101, 9505940301,
               9505940102, 9510940501, 9510940601,
               9510940701, 9510940788, 9512940702,
               9512940703, 9601940901, 9601940902,
               9602941001, 9602941002, 9602941003,
               9606950101, 9606950102, 9606950103,
               9606950201, 9606950202, 9607950203,
               9607950204, 9607950205, 9607950301;
               (b) 9505940301, 9510940601, 9601940901,
               9602941001, 9607950301.
MANUFACTURER   PyMaH Corporation, North Hollywood,
               California.
RECALLED BY    PyMaH Corporation, Flemington, New Jersey, by
               letter sent March 27, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Germany, Spain, Ireland,
               Poland, Turkey.
QUANTITY       677.25 cases; firm estimated that 200 cases
               (64 packages per case) remained on market at
               time of recall initiation.
REASON         Devices were found not to meet labeling claims
               of an 18-month shelf life.

_______________
PRODUCT        Ektachem Clinical Chemistry Slides Digoxin
               (DGXN), for in-vitro diagnostic use to
               quantitatively measure the drug digoxin in
               serum and plasma.  Recall #Z-410-6.
CODE           All codes with expiration date of 1/1/96 or
               earlier.
MANUFACTURER   Johnson & Johnson Clinical Diagnostics, Inc.,
               Rochester, New York.
RECALLED BY    Manufacturer, by telephone July 17 and 18,
               1995, followed by letters of July 17, 1995 and
               August 10, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       14,904 cartridges were distributed.
REASON         The digoxin slides may exhibit a positive bias
               overtime on the first slide from each
               cartridge.

                             -7-_______________
PRODUCT        Caulk Friction Grip Carbide Super Burs, used
               for cutting teeth.  Recall #Z-434-6.
CODE           Lot numbers 941201 through 950228.
MANUFACTURER   Dentsply International Caulk Division, York,
               Pennsylvania.
RECALLED BY    Manufacturer, by letter March 28, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       537,000 were distributed.
REASON         The device is subject to dulling of the
               cutting edges or the bur tip, and possible
               breakage in use at the weld location.

_______________
PRODUCT        Ektachem Clinical Chemistry Slides for
               Cholesterol (CHOL), for in-vitro diagnostic
               use to quantitatively measure cholesterol in
               serum plasma.  Recall #Z-444-6.
CODE           Catalog #168-8290, lot numbers all with an
               expiration date of 7/96 or before.
MANUFACTURER   Johnson & Johnson Clinical Diagnostics, Inc.,
               formerly Eastman Kodak Co., Clinical
               Diagnostics Division, Rochester, New York.
RECALLED BY    Johnson & Johnson Clinical Diagnostics, Inc.,
               Rochester, New York, by letter sent on
               November 29, 1993.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,336,000 cartridges were distributed.
REASON         There was a drift in cholesterol results
               within the same day caused by slide
               temperature and humidity.

_______________
PRODUCT        Kodak Ektachem Clinical Chemistry Slides
               (TRIG), for in-vitro diagnostic use, for the
               quantitative measurement of triglycerides
               concentration in serum or plasma.  
               Recall #Z-445-6.
CODE           Catalog #1648088; all lot numbers with an
               expiration date of 11/96 or earlier.
MANUFACTURER   Johnson & Johnson Clinical Diagnostics,
               formerly Eastman Kodak Company, Clinical
               Diagnostics Division, Rochester, New York.
RECALLED BY    Johnson & Johnson Clinical Diagnostics, Inc.,
               Rochester, New York, by letter sent May 18,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 760,000 cartridges were
               distributed.

                             -8-REASON         Triglyceride slides on-analyzer stability did
               not support a four week expiration claim.

_______________
PRODUCT        OPUS Total B-hCG Test Modules for in-vitro
               diagnostic use for the quantitative
               measurement of intact and free beta subunits
               of human chorionic gonadotropin (hCG) in
               serum, plasma, and urine.  
               Recall #Z-446-6.
CODE           Catalog #304-050, Lot #BGFO EXP 5/18/96.
MANUFACTURER   Behring Diagnostics, Inc., Westwood,
               Massachusetts.
RECALLED BY    Manufacturer, by telephone on January 30,
               1996, followed by letter.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       409 boxes (50 units per box) were distributed.
REASON         Insufficient conjugate in the wells may cause
               false negatives or lower than expected
               results.

_______________
PRODUCT        ILab IL Test (tm) Albumin Reagent, intended
               for the quantitative in-vitro diagnostic
               determination of serum albumin using the ILab
               Chemistry Systems.  Recall #Z-447-6.
CODE           Catalog #182500, Lot #N1133691 (kit), Lot
               #N0833100 (reagent).
MANUFACTURER   Instrumentation Laboratory Company,
               Orangeburg, New York (responsible firm).
RECALLED BY    Instrumentation Laboratory Company, (ILC),
               Lexington, Massachusetts, by letter dated
               January 14, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Canada.
QUANTITY       2 kits were distributed.
REASON         The Albumin reagent had mold growth and
               reagent deterioration.

_______________
PRODUCT        Monarch IL Test Albumin Reagent, for the
               quantitative in-vitro diagnostic determination
               of serum albumin using the IL monarch (r)
               Chemistry Systems.  Recall #Z-448-6.
CODE           Catalog #35191, Lot #N0932200 EXP 11/94.
MANUFACTURER   Instrumentation Laboratory Company, ILC),
               Orangeburg, New York (responsible firm).
RECALLED BY    Instrumentation Laboratory Company, Lexington,
               Massachusetts, by letter sent during January
               1994.  Firm-initiated recall complete.

                             -9-DISTRIBUTION   Florida, Oregon, Texas, Kansas, Louisiana,
               Wisconsin, Missouri, Pennsylvania, Ohio,
               Kentucky, Michigan, Arkansas.
QUANTITY       704 kits were distributed.
REASON         The Albumin reagent had mold growth and
               reagent deterioration.

                            -10-


END OF ENFORCEMENT REPORT FOR FEBRUARY 21, 1996.  BLANK PAGES MAY
FOLLOW.

                                   ####