FDA Home Page | Search FDA Site | FDA A-Z Index | Contact FDA
The FDA Enforcement Report is published weekly by the Food and Drug Administration, Department of Health and Human Services. It contains information on actions taken in connection with agency Regulatory activities.
May 23, 2007
07-21
______________________________
PRODUCT
Russian Sea (Russkoe More) brand Atlantic Herring Jumbo Marinated, Net Wt.
1300 g (45.85 Oz). The product is packaged in an unopened two piece metal can.
The product is labeled mostly in Russian with a white Nutrition Facts Panel
label in English. --- According to New York State Department of Agriculture
and Markets lab report, the fish are whole & 10-12 inches in length with
five per can, UPC bar code 4 605561 000153, Recall # F-260-7
CODE
Embossed into top of can: ''011006 413N74 1P''
RECALLING FIRM/MANUFACTURER
Recalling Firm: Royal International Trading Inc, Brooklyn, NY, by
press release on January 30, 2007.
Manufacturer: Ch. P. Lipina L.A, Kaliningrad, Russia. Firm initiated
recall is complete.
REASON
The product is a processed uneviscerated fish based on sampling & analysis
by New York State Department of Agriculture and Markets. This poses a health
risk if consumed as it may contain Clostridium botulinum spores which can cause
botulism.
VOLUME OF PRODUCT IN COMMERCE
Unknown
DISTRIBUTION
NY
___________________________________
PRODUCT
Product is an Oatmeal Cookie (label does not state product to be an Oatmeal
Cookie). Label includes “Down Unders' nummy nutrition Mobile Meal egg-free.
Product is packaged in 7 oz and 12 oz clear plastic cellophane wrap with an
adhesive label affixed to packaging. Cookies are sold individually; fifty (50)
cookies are placed in a box for delivery to consignees, Recall # F-261-7
CODE
No codes used on product
RECALLING FIRM/MANUFACTURER
Cuisine Med. Catering, Port Townsend, WA, by telephone on March 26, 2007 and
visit on March 27, 2007. Firm initiated recall is complete.
REASON
The product lists whey as an ingredient but does not list milk as the source
of the whey as required by FALCPA (Food Allergen Labeling and Consumer Protection
Act).
VOLUME OF PRODUCT IN COMMERCE
7 oz. – 50 cookies; 12 oz. – 541 cookies
DISTRIBUTION
WA
___________________________________
PRODUCT
Icapeb Rellenos Caramelos Surtti Frutas, Weight 450G/ 15/9 oz. UPC 7 861044
600754, Recall # F-262-7
CODE
No codes used on product
RECALLING FIRM/MANUFACTURER
Recalling Firm: Family Food Distributor, Inc, Kearny, NJ, by telephone
beginning on November 15, 2006.
Manufacturer: Icapeb, Quito, Ecuador. Firm initiated
recall is complete.
REASON
Candies contain undeclared and possibly uncertified colors. Analysis of the
product by the New York State Department of Agriculture and Markets found tartrazine
(certifiable as FD&C Yellow No. 5), Sunset Yellow FCF (certifiable as FD&C
Yellow No. 6), Brilliant Blue FCF (certifiable as FD&C Blue No. 1) and
Allura Red AC (certifiable as FD&C Red No. 40).
VOLUME OF PRODUCT IN COMMERCE
20 cases of 24 x 15.9 oz. bags
DISTRIBUTION
NJ, NY, CT, NC, and OH
___________________________________
PRODUCT
Sweet Dairy Whey (Edible Grade) Product Code: 7012, A Derivative of Milk, Plant
49-66, unit size 50 pound paper bags lined with polyethylene liner, stored
under ambient conditions, Recall # F-263-7
CODE
SM06300, SM06307, SM06296
RECALLING FIRM/MANUFACTURER
Recalling Firm: Protient, Inc, Saint Paul, MN, by letters on November
18, 2006 and November 21, 2006.
Manufacturer: Utah Proteins LLC, Smithfield, UT. Firm initiated
recall is ongoing.
REASON
Sweet Dairy Whey Powder may be contaminated with Salmonella. The product was
used as an ingredient in foods that were further pasteurized.
VOLUME OF PRODUCT IN COMMERCE
97,500 pounds
DISTRIBUTION
NC, MO, and UT
___________________________________
PRODUCT
Driscoll's brand fresh Strawberries in 1-lb clamshell containers (8 per case),
4-lb clamshell containers (2 per case), and 1 pint clamshell containers (12
per case). Recall # F-264-7
CODE
Lot # 0101770717
RECALLING FIRM/MANUFACTURER
Recalling Firm: Driscoll Strawberry Association, Watsonville, CA, by
d-mail and telephone beginning on July 19, 2006.
Manufacturer: Royal Oaks Farms Ranch, Watsonville, CA. Firm initiated
recall is complete.
REASON
The product has been found to contain 2 .65 ppm of Methomyl (lannate) pesticide.
The EPA tolerance for this pesticide in strawberries is 2 ppm.
VOLUME OF PRODUCT IN COMMERCE
3,014 cases
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
LEVIROL Sexual Performance Formula Dietary Supplement, a proprietary blend
of: Epimedium (Epimedium sagittatum leaf), 2 Capsules per pack, Recall # D-795-2007
CODE
Lot # 0606332 expires 06/09 & Lot # 0605143 expires 05/09
RECALLING FIRM/MANUFACTURER
Recalling Firm: Levirol, LLC, Tempe, AZ, by letter on January 28,
2007, and February 12, 2007.
Manufacturer: Bactolac Pharmaceutical, Inc, Hauppauge, NY. Firm
initiated recall is ongoing.
REASON
Unapproved New Drug; product found to contain the undeclared ingredient Aminotadalafil,
an analogue of Tadalafil, the active pharmaceutical ingredient found in an
FDA approved drug used for erectile dysfunction.
VOLUME OF PRODUCT IN COMMERCE
22,658 capsules (11,329 x2 capsule packages)
DISTRIBUTION
Nationwide and Canada
___________________________________
PRODUCT
LIVIRO3 Natural Energy Enhancer Nutritional Supplement, 450mg Capsules, packaged
in boxes of 10 capsules and bottles of 20 capsules, Recall # D-796-2007
CODE
The firm indicated the distributed 5 lots(#1-5) Lot#1 Exp. Date 03/2006, Lot#2
Exp. Date 11/2008, Lot#3 Exp. Date 0/2009, Lot#4 Exp. Date 07/2009 & Lot#5
Exp. Date 07/2009 NOTE: THIS IS WHAT WAS FOUND ON SAMPLE COLLECTED- Underneath
the aluminum try that contains all 10 capsules of LIVIRO3 (found inside box),
there are lot codes and expiration dated imprinted on it. The lot code for
the FDA sample is LOT 0730 EXP 07/30/2009. There might be other lot codes since
it appears that the lot code correspond with the expiration date. Note: Other
codes pending
RECALLING FIRM/MANUFACTURER
Recalling Firm: Ebek, Inc, Los Angeles, CA, by telephone on January
10, 2007, press release on January 19, 2007 and January 29, 2007, and by letters
beginning January 29, 2007.
Manufacturer: West Coast Laboratories Inc, Gardena, CA. Firm
initiated recall is ongoing.
REASON
Unapproved New Drug; product to contain the undeclared ingredient Tadalafil,
the active pharmaceutical ingredient in an FDA approved drug used to treat
erectile dysfunction
VOLUME OF PRODUCT IN COMMERCE
278,260 capsules
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
a) Children's Mucinex Expectorant (Guaifenesin, USP, 100mg/5mL), Grape Flavor
Liquid,
4 fl oz. (118mL) bottle, NDC 63824-173-64, Recall # D-797-2007;
b) Children's Mucinex Cough (Guaifenesin, USP 100mg/5mL and Dextromethorphan
HBr 5mg/5mL), Cherry Flavor Liquid, 4 fl oz. (118mL), Expectorant
and Cough
Suppresent, NDC 63824-174-64, Recall # D-798-2007
CODE
a) Lot #605010, 605011, 605012, 606001, 606002, 606003, 606004, 606005,
606006, 606007, 606008, 606009, 606010, 606011, 606012,
606014,
606015, 606016, 606017, 606018, 606019, 606020, 608004,
608005,
608006, 608007, 608008, 608009, 608010, 608031, 608032,
609001,
609002, 609003, 609004, 609005, 609006, 609023, 609024,
609025,
609026, 609027, 609028, 609029, 609030, 609031, 609032,
609033,
609034, 609035, 610023, 610024, 610025, 610026, 610027,
610028,
610029, 610030, 610032;
b) Lot #605007, 605008, 605009, 605021, 605022, 605023, 605024, 605025,
605026, 605027, 605028, 605029, 607001, 607002, 607003,
607004,
607005, 607006, 607007, 607008, 607009, 607010, 607011,
608011,
608012, 608013, 608014, 608015, 608016, 608017, 608018,
608019,
608020, 608021, 608022, 608028, 609007, 609008, 609009,
609010,
609011, 609012, 609013, 609014, 609015, 609016, 609017,
609018,
609019, 609020, 609021, 609022, 610003, 610004, 610005,
610006,
610007, 610008, 610009, 610010, 610011, 610012, 610013,
610014,
610015, 610016, 610017, 610018, 610019
RECALLING FIRM/MANUFACTURER
Recalling Firm: Adams Respiratory Therapeutics, Inc, Fort Worth, TX,
by letter on April 2, 2007.
Manufacturer: QPharma, LLC, Lawrenceville, NJ. Firm initiated recall
is ongoing.
REASON
Misbranded; calibrated dosage dispensing cup may be confusing to consumers
as measurements are shown in units of teaspoon and tablespoon while dosages
are all listed in teaspoons
VOLUME OF PRODUCT IN COMMERCE
2,102,856 bottles
DISTRIBUTION
Nationwide and PR
________________________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Recall # B-1251-07
CODE
Unit: 041FV66018
RECALLING FIRM/MANUFACTURER
American National Red Cross, Alabama Region, Birmingham, AL, by telephone on
March 11, 2006 and by letter dated March 14, 2006. Firm initiated recall
is complete.
REASON
Blood product, collected from a donor who was at risk for Leishmaniasis, was
distributed
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
AL
___________________________________
PRODUCT
a) GE LightSpeed VCT / Light Speed VCT standard configuration, Model 5124069,
CT Scanner System, Recall # Z-0802-2007;
b) GE LightSpeed VCT Select / LightSpeed Pro32, Model 5126093,
CT Scanner system, Recall # Z-0803-2007;
c) GE LightSpeed Pro 16 (100kW/80kW) Models 2357739, 2357739-3,
2357739-4, 2357739-6, 2357739-7 and 2357739-8, CT Scanner
systems,
Recall # Z-0804-2007;
d) GE LightSpeed Xtra / GE LightSpeed RT Pro 16, Model 2374681-6,
CT Scanner systems, Recall # Z-0805-2007;
e) GE Discovery VCT, Volume PET-CT scanner, Model 5124069-3,
Recall # Z-0806-2007
CODE
All serial numbers
RECALLING FIRM/MANUFACTURER
General Electric Med Systems, LLC, Waukesha, WI, by letter dated December 4,
2006 beginning December 21, 2006. Firm initiated recall is ongoing.
REASON
Oil may leak onto the port window of the CT tube and may result in image artifacts
that could lead to misdiagnosis.
VOLUME OF PRODUCT IN COMMERCE
1,817 devices
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
Roche PTH (parathyroid hormone test system) for use in conjunction with Roche
cobas e 411, cobas e 601, Elecsys 1010, Elecsys 2010 and Modular E 170 analyzers;
catalog no. 11972103122, Recall # Z-0810-2007
CODE
Lots 17723501, exp. 1/31/08; 17723502, exp. 1/31/08; 17642601, exp. 11/30/07;
17642602, exp. 11/30/07; 17450901, exp. 5/31/07; 17450902, exp. 5/31/07 and
17450903, exp. 5/31/07
RECALLING FIRM/MANUFACTURER
Recalling Firm: Roche Diagnostics Corp., Indianapolis, IN, by letter
dated April 19, 2007.
Manufacturer: Roche Diagnostics Gmbh, Mannheim, Germany. Firm initiated
recall is ongoing.
REASON
Negative bias caused by hemolysis interference at the low end of the reference
range could cause erroneous diagnosis of hypoparathyroidism.
VOLUME OF PRODUCT IN COMMERCE
14,503 systems
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
STA Compact and STA Compact CT For in vitro Diagnostic Use Only In vitro diagnostic
analyzer, Recall # Z-0812-2007
CODE
All units with the last 4 digits of the serial number under 3450. Serial numbers
5050012 6060290 6070319 6070346 6090389 6090416 6090417 6090418 6100419 7010600
7010601 7010602 7010603 7010604 7010605 7010606 7010607 7010608 7010609 7010611
7010612 7020648 7020649 7020650 7020651 7020652 7020653 7020654 7020655 7020656
7020683 7020684 7020685 7020686 7020687 7030745 7030746 7030747 7030748 7030749
7030751 7030752 7040808 7040809 7040810 7040811 7040812 7040813 7040814 7040815
7040816 7040817 7050834 7050835 7050836 7050837 7050838 7050871 7050872 7050873
7050874 7050875 7050876 7050877 7050878 7050879 7060931 7060932 7060933 7060935
7060936 7060937 7060938 7060939 7060940 7091037 7091038 7091039 7091040 7091041
7091060 7091079 7091080 7091081 7121199 7121200 7121201 7121202 7121203 8011222
8011223 8011224 8011225 8011226 8011228 8011229 8011230 8011231 8021232 8021233
8021234 8021235 8021236 8021237 8021238 8021239 8021240 8021241 8021242 8021243
8021244 8021245 8021246 8021247 8021248 8021249 8021250 8021251 8031259 8031260
8031261 8031262 8031263 8031264 8031265 8031269 8031286 8031287 8031288 8031289
8031290 8041304 8041305 8041306 8041307 8041308 8041309 8041310 8041312 8041313
8041314 8041315 8041316 8041317 8041318 8061357 8061358 8071387 8071388 8071389
8071390 8071391 8071392 8071393 8071394 9011495 9011496 9011497 9011498 9011499
9011500 9011501 9021521 9021524 9021525 9021526 9021528 9021529 9021530 9021531
9021532 9021537 9021538 9021539 9021540 9031557 9031558 9031559 9041594 9041595
9041596 9041597 9041598 9041599 9051600 9051601 9051602 9051603 9051604 9061645
9061646 9061647 9061648 9061659 9061660 9061661 9061662 9061663 9071689 9071690
9071691 9071693 9071694 9071695 9071696 9071700 9081714 9081715 9081729 9081730
9081731 9081736 9081737 9091739 9091741 9091768 9091769 9091770 9091774 9091775
9091776 9091777 9091778 9091779 9091783 9091784 9091785 9091786 9091787 9101842
9101843 9101844 9101845 9101846 9101847 9101848 9101856 9111858 9111861 9111862
9111863 9111864 9111865 9111866 9111867 9111868 9111874 9111876 9111877 9111878
9111879 9111880 9111881 9111882 9111906 9111907 9111908 9111909 9111910 9111913
9111914 9111916 9111917 9111918 9111919 9111920 9121937 9121939 9121940 9121942
9121943 9121945 9121948 9121949 9121950 9121951 11971 11972 11973 11974 11975
11976 11977 11991 11992 11993 11994 11995 11996 11997 11998 21999 22000 22003
22008 22009 22010 22014 22015 22016 22017 22018 22019 22023 32036 32037 32038
32039 32040 32041 32042 32061 32062 32063 32064 32065 32071 32073 32074 32083
42109 42110 42111 42112 52141 52142 72239 72240 72241 82250 82251 82254 82255
82256 82257 82258 82259 82266 82267 82268 92269 92277 92278 92280 92281 92283
92288 92289 92290 92291 92293 92294 102299 102301 102302 102303 102304 102328
102331 102333 102334 102335 102336 102338 102339 102340 102342 102344 102345
102346 112350 112351 112352 112353 112354 112355 112356 112357 112368 112371
112372 112379 112380 112381 112382 112383 112384 112385 112386 112387 112388
122423 122424 122425 122427 122428 122429 122430 122431 122432 122433 1012434
1012435 1012456 1012457 1012458 1012460 1012461 1012462 1012463 1012465 1012474
1012475 1012476 1022502 1022513 1022514 1022515 1022516 1022517 1022518 1022520
1022521 1032523 1032524 1032527 1032528 1042578 1042579 1042580 1042581 1042582
1042583 1042584 1042585 1042586 1042587 1042588 1042589 1042590 1042591 1052634
1052635 1052636 1052637 1052638 1052639 1052640 1052641 1052642 1052643 1052645
1052647 1062676 10626771062678 1062683 1062684 1062685 1062686 1062687 1062688
1062689 1062690 1062691 1062692 1062693 1062694 1062696 1062697 1062699 1062700
1062702 1062703 1062704 1062705 1062706 1072751 1072752 1072753 1072754 1072755
1072757 1072758 1072759 1072766 1072767 1072768 1072769 1072770 1092815 1092816
1092817 1092818 1092819 1092820 1092821 1092822 1092823 1092824 1092825 1092826
1092827 1092828 1092830 1092831 1092832 1092833 1092834 1092836 1092848 1092849
1092850 1092851 1092852 1092853 1092854 1092855 1092856 1092857 1092858 1092861
1092862 1102869 1102871 1102873 1102874 1102875 1102876 1102877 1102895 1102911
1112927 1112949 1112950 1112951 1112952 1112953 1112954 1112955 1112956 1112957
1112958 1112959 1112960 1112961 1112962 1123008 1123009 1123011 1123013 1123014
1123015 1123016 2013050 2013051 2013052 2023108 2023109 2023110 2023111 2023112
2023113 2023114 2023115 2023116 2023117 2023118 2023121 2023122 2023123 2023124
2023125 2023126 2023127 2023128 2023129 2023130 2023131 2023132 2023133 2023134
2023151 2023152 2023153 2023188 2023189 2023190 2023191 2023192 2033193 2033194
2033195 2033196 2033197 2033198 2033199 2033200 2033201 2033202 2033203 2043224
2043225 2043226 2043227 2043228 2043229 2043230 2043231 2043232 2043233 2043234
2043237 2043238 2043239 2043240 2043241 2043242 2043243 2043244 2043245 2043251
2043252 2043253 2043254 2043255 2043256 2043257 2043258 2043259 2043260 2043266
2043268 2043269 2043271 2043272 2043288 2043289 2043291 2043292 2043293 2043295
2043296 2043297 2053299 2053300 2053301 2053302 2053303 2053304 2053305 2053346
2053348 2053349 2053350 2053351 2053352 2053353 2053354 2053355 2053356 2053357
2053358 2053359 2053360 2053361 2053362 2063364 2063365 2063366 2063367 2063368
2073397 2073398 2073399 2073400 2073401 2073402 2073404 2073405 2073406 2073418
2073419 2073420 2073421 2073422 2073423 2073424 2073425 2073446 2073447 2073448
2073449
RECALLING FIRM/MANUFACTURER
Recalling Firm: Diagnostica Stago, Inc., Parsippany, NJ, by representative
visit beginning on January 18, 2007.
Manufacturer: Diagnostica Stago, Franconville, France. Firm initiated
recall is ongoing.
REASON
Result of complaints of field failures of the silicon bridge rectifier, resulting
in smoke billowing out of the unit and burnt circuit boards, wires and connectors.
VOLUME OF PRODUCT IN COMMERCE
693 units
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
BacT/VIEW® Software, all versions. in vitro diagnostic database software,
Recall # Z-0814-2007
CODE
Software Version C.30a -22105-68, Update Disk; 22105-69 Install Disk
RECALLING FIRM/MANUFACTURER
bioMerieux, Inc, Durham, NC, by customer notification on December 19, 2006. Firm
initiated recall is ongoing.
REASON
BacT/VIEW will be unable to run scheduled macro applications unless the user
touches the BacT/VIEW touch screen or clicks a function after resynchronizing
the clocks.
VOLUME OF PRODUCT IN COMMERCE
1233 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
GE Healthcare-Americas Signa 3.0T EXCITE Whole Body Magnetic Resonance System,
Recall # Z-0815-2007
CODE
00000249264MR3, 00000257756MR7, 00000255196MR8, 00000246630MR8, 00000259107MR1,
00000248178MR6, 00000257451MR5, 00000256200MR7, 00000260871MR9, 00000252451MR0,
00000259108MR9, 00000254480MR7, 00000254482MR3, 00000256198MR3, 00000249263MR5,
00000250755MR6, 00000251090MR7, 00000GE0131001, 00000GE0213040, 00000254375MR9,
00000GE0205028, 00000GE0210034, 00000227995MR8, 00000GE0212037, 00000GE0226049,
00000GE0410083, 00000260871MR9, 00000GE0234057, 00000GE0123041, 00000GE0302068
RECALLING FIRM/MANUFACTURER
GE Medical Systems LLC, Waukesha, WI, by visit beginning June, 2006. Firm
initiated recall is ongoing.
REASON
To provide the 3.0T MRI systems users with the 8-channel brain coil the proper
coil weighting software. The proper coil weighting factors for the eight-channel
brain coil were not included in the software.
VOLUME OF PRODUCT IN COMMERCE
30 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
Cranex Tome panoramic x-ray unit with Spiral Tomography for cross-sectional
imaging, Recall # Z-0816-2007
CODE
Serial numbers X51146, X51147, X51148, X51154, X51155, X51156, X51170, X51171,
X61188
RECALLING FIRM/MANUFACTURER
Recalling Firm: Instrumentarium Dental, Inc, Milwaukee, WI, by letter
dated April 28, 2006.
Manufacturer: Soredex Palodex Group Oy, Tuusula, Finland. Firm
initiated recall is ongoing.
REASON
There may be a material defect in the up/down movement axle of the unit. This
may cause the upper part of the unit to suddenly come down.
VOLUME OF PRODUCT IN COMMERCE
9 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
Certain MicroMiniplant Straight Healing Abutment. Dental implant. Model numbers:
ISMHA32, ISMHA33, ISMHA34 and ISMHA36, Recall # Z-0811-2007
CODE
Lot numbers: 593025-5, 593027-5, 598408-5, 593028-5, 599866-5 and 593029-5
RECALLING FIRM/MANUFACTURER
Biomet 3i, Palm Beach Gardens, FL, by telephone and letter on April 25, 2007. Firm
initiated recall is ongoing.
REASON
The Straight Healing Abutment did not fully seat to the restorative platform
of the implant, leaving a gap between the abutment and the implant.
VOLUME OF PRODUCT IN COMMERCE
362 units
DISTRIBUTION
Nationwide and Internationally
END OF ENFORCEMENT REPORT FOR MAY 23, 2007
###