RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        DAPA brand Extra Strength Acetaminophen Capsules, all size
               packages, OTC.  Recall #D-011-6.
CODE           All lots.
MANUFACTURER   Ferndale Laboratories, Inc., Ferndale, Michigan.
RECALLED BY    Manufacturer, by letter dated February 14, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       4,900,000 capsules were distributed.
REASON         Product does not meet dissolution specifications.

_______________
PRODUCT        Physician samples of Pramasone (Pramoxine Hydrochloride and
               Hydrocortisone Acetate) Lotion 1%, 2.5%, 7.5 ml, in 15 ml
               bottle, Rx indicated for topical relief of the inflammatory
               and pruritic manifestations of corticosteroid-responsive
               dermatosis.  Recall #D-012-6.
CODE           Physician sample lot numbers:  54-600, 53-234, 53-232.
MANUFACTURER   Ferndale Laboratories, Inc., Ferndale, Michigan.
RECALLED BY    Manufacturer, by letter May 22, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       308,995 bottles were distributed
REASON         Subpotency of the pramoxine component.
RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  
               Recall #B-037/038-6.
CODE           Unit 28J21081.
MANUFACTURER   American Red Cross Blood Services, Savannah, Georgia.
RECALLED BY    Manufacturer, by letters dated March 24, 1995 and May 12,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Georgia and Switzerland.
QUANTITY       1 unit of each component.
REASON         Blood products, which were not tested for antibody to the
               hepatitis C virus encoded antigen (anti-HCV), were
               distributed.

_______________
PRODUCT        (a) Whole Blood CPDA-1; (b) Red Blood Cells.
               Recall #B-041/042-6.
CODE           Unit numbers:  (a) DD00806; (b) 131255.
MANUFACTURER   Blood Bank of Alaska, Anchorage, Alaska.
RECALLED BY    Manufacturer, by letter dated January 24, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Alaska. 
QUANTITY       1 unit of each component.
REASON         Blood products, which were collected from a donor who had a
               medical history of cancer, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
               Recall #B-046/048-6.
CODE           Unit #128386.
MANUFACTURER   Blood Bank of Alaska, Anchorage, Alaska.
RECALLED BY    Manufacturer, by letter dated August 2, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Alaska and Florida.
QUANTITY       1 unit of each component.
REASON         Blood products, which were collected from a donor with a
               history of IV drug use, were distributed.

_______________
PRODUCT        Platelets.  Recall #B-054-6.
CODE           Unit #28105-8168.
MANUFACTURER   Blood Systems, Inc., doing business as United Blood
               Services, San Angelo, Texas.
RECALLED BY    Manufacturer, by telephone on May 2, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product, which was prepared from a whole blood unit
               with an extended collection time, was distributed.


                                    -2-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Platelets.  Recall #B-040-6.
CODE           Unit #27C82850.
MANUFACTURER   American National Red Cross, Johnstown, Pennsylvania.
RECALLED BY    Manufacturer, by letter dated September 3, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit.
REASON         Blood product, which tested positive for the anti-P1
               antibody but labeled as negative for unexpected
               alloantibodies, was distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-043-6.
CODE           Unit #131255.
MANUFACTURER   Blood Bank of Alaska, Anchorage, Alaska.
RECALLED BY    Manufacturer, by letter dated January 24, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product, which was collected from a donor who had a
               medical history of cancer, was distributed.


RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II
========================
_______________
PRODUCT        APEX brand Universal Irrigation System One-Connection
               Arthroscopic Tubing Sets, Model C7120, indicated for joint
               distention.  Recall #Z-069-6.
CODE           Lot numbers:  29989-30036.
MANUFACTURER   Zimmer Patient Care Division, Dover, Ohio.
RECALLED BY    Linvatec, Inc., Largo, Florida, by letter September 8, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       19,144 units were distributed.
REASON         There is a leak at the isolator valve at the junction of the
               small tubing line and isolator cap.

_______________
PRODUCT        Baxter Surgical Group Pouched Procedure Trays.
               Recall #Z-079/110-6.
               1.  Catalog No. 4547, Baxter Pharmaseal Genetic
               Amniocentesis Tray with 22 Ga x 3-1/2" Spinal Needle with
               Depth Calibrations at 1 cm  Intervals;
               2.  Catalog No. BAK2015, Baxter Pharmaseal Illinois
               Sternal/Iliac Bone Marrow Aspiration Tray with 15 Ga
               Aspiration Needle (Adjustable 3/8" - 1-7/8");
               3.  Catalog No. BAK2018, Baxter Pharmaseal Illinois
               Sternal/Iliac Bone Marrow Aspiration Tray with 18 Ga
               Aspiration Needle (Adjustable Length 1/16" - 1-7/16");

                                    -3-               4. Catalog No. BAK3353, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy Tray with 13 Ga, 3-1/2" Biopsy/Aspiration
               Needle;
               5.  Catalog No. BAK3411, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy Tray with 11 Ga, 4" Biopsy/Aspiration Needle;
               6.  Catalog No. BAK3583, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy and Aspiration Tray with 13 Ga, 3-1/2"
               Biopsy/Aspiration Needle and 18 Ga Aspiration Needle
               (Adjustable Length 1/16" - 1-7/16");
               7.  Catalog No. BAK4508, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy and Aspiration Tray with 8 Ga, 4"
               Biopsy/Aspiration Needle and 18 Ga Aspiration Needle
               (Adjustable Length 3/8" - 1-7/8");
               8.  Catalog No. BAK4511, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy and Aspiration Tray with 11 Ga, 4"
               Biopsy/Aspiration Needle and 15 Ga Aspiration Needle
               (Adjustable Length 3/8" - 1-7/8");
               9.   Catalog No. BEK2015, Baxter Pharmaseal Economy Illinois
               Sternal/Iliac Bone Marrow Aspiration Tray with 15 Ga
               Aspiration Needle (Adjustable 3/8" - 1-7/8");
               10.  Catalog No. BEK3411, Baxter Pharmaseal Economy Jamshidi
               Bone Marrow Biopsy Tray with 11 Ga, 4" Biopsy/Aspiration
               Needle;
               11.  Catalog No. BEK4511, Baxter Pharmaseal Economy Jamshidi
               Bone Marrow Biopsy and Aspiration Tray with 11 Ga, 4"
               Biopsy/Aspiration Needle and 15 Ga Aspiration Needle
               (Adjustable Length 3/8" - 1-7/8");
               12.  Catalog No. BK1000, Baxter Pharmaseal Bone Marrow
               Biopsy Tray Without Needle; 
               13.   Catalog No. CAK3411, Baxter Pharmaseal Contoured
               Jamshidi Bone Marrow Biopsy Tray with 11 Ga, 4" Bone Marrow
               Aspiration Needle;
               14.  Catalog No. CAK4511, Baxter Pharmaseal Contoured
               Jamshidi Bone Marrow Biopsy Tray with 11 Ga, 4" Aspiration
               Needle and 15 Ga Illinois Sternal/Iliac Aspiration Needle
               (3/8" - 1-7/8");
               15.  Catalog No. EJT3353, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy Tray with 13 Ga, 3-1/2" Ergonomically Designed
               Biopsy/Aspiration Needle;
               16.  Catalog No. EJT3411, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy Tray with 11 Ga, 4" Ergonomically Designed
               Biopsy/Aspiration Needle;
               17. Catalog EJT3583, Baxter Pharmaseal Jamshidi Bone Marrow
               Biopsy and Aspiration Tray with 13 Ga, 3-1/2" Ergonomically
               Designed Biopsy/Aspiration Needle and 18 Ga Aspiration
               Needle (Adjustable 1/16" - 1-7/16");
               18. Catalog No. EJT4508, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy and Aspiration Tray with 8 Ga, 4"
               Ergonomically Designed Biopsy/Aspiration Needle and 15 Ga
               Aspiration Needle (Adjustable 3/8" - 1-7/8");

                                    -4-               19. Catalog No. EJT4511, Baxter Pharmaseal Jamshidi Bone
               Marrow Biopsy and Aspiration Tray with 11 Ga, 4"
               Ergonomically Designed Biopsy/Aspiration Needle and 15 Ga
               Aspiration Needle (Adjustable 3/8" - 1-7/8");
               20.  Catalog No. LNT9000, Baxter Pharmaseal Lazarus-Nelson
               Peritoneal Lavage Tray;
               21.  Catalog No. PC-1000, Baxter Pharmaseal Paracentesis
               Tray with Boasberg Catheter Set;
               22. Catalog No. PTC6022A, Baxter Pharmaseal Jamshidi
               Percutaneous Transhepatic Cholangiography Procedure Tray
               with 22 Ga X 6" Skinny-Chiba Needle with Centimeter Depth
               Markings;
               23.  Catalog No. PTC8022A, Baxter Pharmaseal Jamshidi
               Percutaneous Transhepatic Cholangiography Procedure Tray
               with 22 Ga X 8" Skinny-Chiba Needle with Centimeter Depth
               Markings;
               24.  Catalog No. SK1015A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 15 Ga (100 mm x 1.9 mm)
               Needle/Syringe with Centimeter Depth Markings;
               25. Catalog No. SK1016A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 16 Ga (100 mm x 1.6 mm)
               Needle/Syringe with Centimeter Depth Markings;
               26. Catalog No. SK1017A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 17 Ga (100 mm x 1.4 mm)
               Needle/Syringe with Centimeter Depth Markings;
               27. Catalog No. SK1018A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 18 Ga (100 mm x 1.2 mm)
               Needle/Syringe with Centimeter Depth Markings;
               28.  Catalog No. SK7015A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 15 Ga (70 mm x 1.9 mm)
               Needle/Syringe with Centimeter Depth Markings;
               29. Catalog No. SK7016A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 16 Ga (70 mm x 1.6 mm)
               Needle/Syringe with Centimeter Depth Markings;
               30. Catalog No. SK7017A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 17 Ga (70 mm x 1.4 mm)
               Needle/Syringe with Centimeter Depth Markings;
               31.  Catalog No. SK7018A, Baxter Pharmaseal Jamshidi Soft
               Tissue Biopsy Tray with 18 Ga (70 mm x 1.2 mm)
               Needle/Syringe with Centimeter Depth Markings;
               32. Catalog No OB-1000, Baxter Circumcision Tray;
CODE           All lots beginning with L4 and L5, excluding lots L5K019,
               L5K024, L5K025, L5K023, L5K043, L5K022, L5K040, L5K041,
               L5K042, L5K060, L5K026, L5J058.   
MANUFACTURER   Baxter Healthcare Corporation, Mannford, Oklahoma.
RECALLED BY    Baxter Healthcare Corporation, McGaw Park, Illinois, by
               letters dated September 22, 1995, and October 2, 1995, and
               by Federal Express October 10, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       254,546 tray were distributed; firm estimated that 25
               percent of the product remained on market at time of recall
               initiation.
                                    -5-REASON         The sterility of the device has been compromised due to loss
               of package seal integrity.

_______________
PRODUCT        Medlite and MedAlex Laser Systems, used in dermatology.
               Recall #Z-111/112-6.
CODE           Undetermined.
MANUFACTURER   Continuum Biomedical, Inc., Livermore, California.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action
               plan October 12, 1995.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       361 Medlite units and 3 MedAlex units were distributed.
REASON         Device failed to comply with 21 CFR 1040.11(a)(2) in that
               the operator's manual lacked adequate calibration
               procedures.

_______________
PRODUCT        Nd:YLF Ophthalmic Laser System, Model 2001, used in
               ophthalmic surgery.  Recall #Z-113-6.
CODE           Model 2001.
MANUFACTURER   Intelligent Surgical Lasers, Inc., San Diego, California.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action
               plant October 12, 1995.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       24 units.
REASON         Device failed to comply with 21 CFR 1040.11(a)(2) in that
               the operator's manual lacked adequate calibration
               procedures.

_______________
PRODUCT        LB-6L Bicarb Concentrate, in gallon containers.
               Recall #Z-116-6.
CODE           117H530.
MANUFACTURER   Renal Systems, Division of Minntech Corporation,
               Minneapolis, Minnesota.
RECALLED BY    Manufacturer, by telephone and by fax on or about October 4,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Maryland, Missouri, Mississippi, North Carolina, New Jersey,
               New York.
QUANTITY       1,223 cases (4 1-gallon bottles per case) were distributed.
REASON         The pH of 8.5 exceeds the specification of 8.3.

_______________
PRODUCT        3M Sarns D4 Aortic Arch Cannula with Luer:
               (a) Catalog No. 165264, 3M Part No. 98-0702-0317-3
               (b) Catalog No. 165265, 3M Part No. 98-0702-0318-1
               (c) Catalog No. 165266, 3M Part No. 98-0702-0319-9
               (d) Catalog No. 165267, 3M Part No. 98-0702-0320-7
               (e) Catalog No. 165268, 3M Part No. 98-0702-0321-5
               (f) Catalog No. 165269, 3M Part No. 98-0702-0322-3

                                    -6-               (g) Custom, 3M Part No. 98-0702-0907-1
               (h) Custom, 3M Part No. 98-0702-0946-9
               (i) Custom, 3M Part No. 98-0702-0950-1.  
               Recall #Z-117/125-6.
CODE           All lot numbers.
MANUFACTURER   Sarns 3M Health Care, Ann Arbor, Michigan.
RECALLED BY    Manufacturer, by visits held between August 1 and 12, 1994,
               and by letter dated July 27, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       70,740 units were distributed. 
REASON         A potential packaging defect which may compromise the
               sterility of the product.

_______________
PRODUCT        DVI Rotating Hemostatic Valve, Model #CA3100, indicated for
               maintaining a fluid-tight seal around a catheter during
               percutaneous atherectomy or angioplasty.  Recall #Z-126-6.
CODE           Lot numbers 507514 and 507515.
MANUFACTURER   Devices for Vascular Intervention, Temecula, California.
RECALLED BY    Manufacturer, by letter sent June 26, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and France.
QUANTITY       986 units were distributed.
REASON         A missing weld causes valve to separate during use.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
=====================
_______________
PRODUCT        Kodak RP XOMAT Developer and Replenisher, acetic acid
               solution, 2.68 gallons/10.1 liters.  Recall #Z-067-6.
CODE           Catalog #1249259, Lot Batch numbers:  9515, 9516, 9517,
               9518.  These kits may contain the suspect Part C, Batch
               #9515. 
MANUFACTURER   Eastman Kodak Company, Rochester, New York.
RECALLED BY    Manufacturer, by letters dated August 3 and 9, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       19,458 units were distributed.
REASON         The Part "C" solution has the potential to cause a drop in
               speed and contrast. 

_______________
PRODUCT        Abiomed BVS 36 Fr. Bendable Atrial Cannula, intended for use
               with the BVS-500 Bi-Ventricular support System.  
               Recall #Z-127-6.
CODE           Catalog #0506-0070, Control numbers:  D9505060170P01 to P12
               and E9505060170P01 to P28.
MANUFACTURER   Abiomed, Inc., Danvers, Massachusetts.
RECALLED BY    Manufacturer, by letter August 23, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.

                                    -7-QUANTITY       39 individual packages were distributed.
REASON         The instructions for use for the cannula were inadvertently
               omitted from the cannula packages.

_______________
PRODUCT        Lifepak 11 Diagnostic Cardiac Monitor with 12 Lead ECG
               Cable, designed for pre-hospital use to detect early ECG
               indicators of AMI.  Recall #Z-128-6.
CODE           All serial numbers.
MANUFACTURER   Physio Control Corporation, Redmond, Washington.
RECALLED BY    Manufacturer, by letter issued on September 19, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       849 cables were distributed.
REASON         There is a potential for the 12 lead ECG cable to break.

                                    -8-


END OF ENFORCEMENT REPORT FOR NOVEMBER 8, 1995.  BLANK PAGES MAY