11/01/1995

 ENFORCEMENT REPORT FOR 11/1/95 


November 1, 1995                                                 95-44

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Various Dinnerware:  (a) Garden Basket Pattern 16-piece
               Stoneware Dinnerware Set; Midnight Satin Pattern: (b) 26-
               piece dinnerware set; (c) Tea Pot/Coffee Server; (d) 6-piece
               Condiment Set; (e) Gravy Boat; (f) 14" Oval Platter; (g) Cup
               and Saucer Set; (h) Orchard Splendor Pattern 24 piece
               dinnerware Set.  Recall #F-026/033-6.
CODE           None.
MANUFACTURER   Shantou Printing Factory, Guangdong Province, China.
RECALLED BY    McCrory Corporation, York, Pennsylvania, by electronic mail
               on May 26, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       36,105 sets were distributed.
REASON         Products contain excessive levels of leachable lead.

_______________
PRODUCT        Wheat bread, sometimes labeled whole wheat bread, 22-1/2
               ounce loaves, in plastic bags, produced under the following
               labels:  Langendorf 100% Wheat Bread, QFC Quality Food
               Centers 100% Wheat Bread; Stock Market Foods 100% Wheat;
               Cascade Mills 100% Whole Wheat Bread; Western Family 100%
               Wheat Bread.  Recall #F-041/042-6.
CODE           "SEP 27" or "27" ink stamped on green plastic bag closures.
MANUFACTURER   Gai's Northwest Bakeries, Langendorf Plant, Seattle,
               Washington.
RECALLED       Gai's Northwest Bakeries Company, Seattle, Washington, by
               visit September 20, 1995, followed by letter October 4,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Washington State.
QUANTITY       3,363 loaves were distributed.
REASON         Product was contaminated with glass, plastic resin, and
               metal.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Breaded Shrimp:  (a) Oriental Breaded Mini Shrimp, Item
               #010559; (b) Breaded Shrimp IQF, Item #075630; (c) Imitation
               Breaded Pouch Pack Shrimp IQF, Item #011726; (d) Breaded
               Mini Shrimp IQF, Item #129513; (e) Imitation Breaded Shrimp
               IQF Shrimp Pouch, Item #086726.  Recall #F-035/039-6.
CODE           Lot numbers:  (a) 344604A; (b) 346450A; (c) 342941A; (d)
               343250A; (e) 342942A.
MANUFACTURER   Bee Gee Shrimp Company, Lakeland, Florida.
RECALLED BY    Manufacturer, by telephone August 21, 1995, followed by
               letter.  Firm-initiated recall ongoing.
DISTRIBUTION   Southeastern and Midwestern United States.
QUANTITY       6,153 cases were distributed; firm estimated that 620 cases
               remained on market at time of recall initiation.
REASON         Product is adulterated due to decomposition.

_______________
PRODUCT        Golden Grain Mission Radiatore Enriched Macaroni Product,
               Net weight 22 ounces.  Recall #F-040-6.
CODE           Bag code:  FEB 21 97 S14 and FEB 21 97 S24; Case code:  JUL
               23 95 S1 and JUL 23 95 S2.
MANUFACTURER   Golden Grain Company, Seattle, Washington.
RECALLED BY    The Quaker Oats Company, Chicago, Illinois, by letters dated
               October 5, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Utah, Washington State.
QUANTITY       1,030 cases were distributed; firm estimated that 10 percent
               of product remained on market at time of recall initiation.
REASON         The radiatore pasta was packaged erroneously in poly bags
               labeled Golden Grain Rigatoni.

_______________
UPDATE         Recall #F-007-6, Strawberry Daiquiri Cocktail Mix, which
               appeared in the October 11, 1995 Enforcement Report is a
               Class III recall.  The reason for recall should read:
               REASON:  The product contains undeclared FD&C Red No. 40.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS I
======================
______________
PRODUCT        (a) Red Blood Cells; (b) Source Leukocytes.
               Recall #B-615/616-5.
CODE           Unit numbers:  60GQ01310 and 60GQ01320.

                                    -2-MANUFACTURER   American National Red Cross, Miami, Florida.
RECALLED BY    Manufacturer, by letters dated October 18, 1993, November
               18, 1993, December 20, 1993, February 25, 1994, March 3 and
               17, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Florida.
QUANTITY       2 units of each component.
REASON         Blood products, which tested repeatedly reactive for the
               antibody to the human T-lymphotropic virus type I (anti-
               HTLV-1) and confirmed positive by unlicensed Western Blot,
               were distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Source Leukocytes.
               Recall #B-617/618-5.
CODE           Unit numbers:  (a) 60FG00115, 60FN01786; (b) 60GS01944,
               60GM00297, 60GM00296, 60FW00317, 60FG00002, 60GQ01559,
               60FG00013, 60GM01866, 60GX01017, 60GM00808.
MANUFACTURER   American National Red Cross, Miami, Florida.
RECALLED BY    Manufacturer, by letters dated October 18, 1993, November
               18, 1993, December 20, 1993, February 25, 1994, March 3 and
               17, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Florida.
QUANTITY       (a) 2 units; (b) 10 units.
REASON         Blood products, tested initially reactive for the antibody
               to the human immunodeficiency virus types 1 and 2 (anti-HIV-
               1 and 2); tested initially reactive for the hepatitis B
               surface antigen (HBsAg); tested repeatedly reactive for the
               antibody to the hepatitis B core antigen (anti-HBc); tested
               repeatedly reactive for the antibody the hepatitis C virus
               encoded antigen (anti-HCV); tested repeatedly reactive for
               syphilis by the Olympus PK7100 Treponema pallidum (TP) test
               and were confirmed positive by the Fluorescent Treponemal
               Antibody Absorption assay (FTA); and were prepared greater
               than eight hours after phlebotomy, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma; (c) Recovered
               Plasma, Pooled.  Recall #B-009/011-6.
CODE           (a) 31H93061, 31H20461; (b) 31G98745, 31S81508, 31H93061,
               31H20461; (c) 86RB2810-1, 88RB2107-1.
MANUFACTURER   American Red Cross, Buffalo, New York.
RECALLED BY    Manufacturer, by letters dated between April 24, 1995 and
               June 7, 1995.  Firm-initiated recall complete.
DISTRIBUTION   New York, Pennsylvania, California, Massachusetts,
               Switzerland.
QUANTITY       (a) 2 units; (b) 4 units; (c) 2 units.
REASON         Blood products which:  tested repeat reactive (RR) for
               HBsAg; were collected from a donor who previously tested RR
               for HBsAg; and were collected from a donor with a medical
               history of hepatitis, were distributed.

                                    -3-_______________
PRODUCT        Red Blood Cells; (b) Recovered Plasma.  Recall #B-013/014-6.
CODE           Unit #47J33350.
MANUFACTURER   American Red Cross Blood Services, Galesburg, Illinois.
RECALLED BY    American Red Cross Blood Services, Peoria, Illinois, by
               letter dated May 2 and 30, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Iowa, California.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from a donor who previously tested
               repeatedly reactive for antibody to immunodeficiency virus
               type 1 (anti-HIV-1) and was not reentered properly, were
               distributed.

_______________
PRODUCT        Platelets, Pooled.  Recall #B-015-6.
CODE           Unit #PP29443.
MANUFACTURER   United Blood Services, Chicago, Illinois.
RECALLED BY    Manufacturer, by telephone on May 4, 1995, followed by
               letter dated May 8, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit.
REASON         Blood product, which was collected from a donor labeled with
               an incorrect expiration date, was distributed.

_______________
PRODUCT        Platelets.  Recall #B-016-6.
CODE           Unit #1156294.
MANUFACTURER   Central Kentucky Blood Center, Inc., Lexington, Kentucky.
RECALLED BY    Manufacturer, by letter dated August 3, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Kentucky.
QUANTITY       1 unit.
REASON         Blood product, which contained anti-Kell antibodies but was
               labeled as negative for unexpected antibodies, was
               distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  
               Recall #B-017/018-6.
CODE           Unit #5F4785.
MANUFACTURER   Puget Sound Blood Center, Seattle, Washington.
RECALLED BY    Manufacturer, by letters dated July 6 and 7, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Washington state.
QUANTITY       1 unit of each component.
REASON         Blood products, which were collected from a donor who had
               close contact with an individual diagnosed with the
               hepatitis C virus encoded antigen (HCV), were distributed.

                                    -4-_______________
PRODUCT        Red Blood Cells Leukocytes Removed.  Recall #B-019-6.
CODE           Unit numbers:  53F89420, 53FH37486, 53FH37487, 53FH37488,
               53FH37489.
MANUFACTURER   American National Red Cross, Baltimore, Maryland.
RECALLED BY    Manufacturer, by letters dated October 3 and 17, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Maryland.
QUANTITY       5 units.
REASON         Blood products which were prepared greater than eight hours
               after phlebotomy, or were prepared from whole blood that was
               stored at room temperature greater than eight hours, were
               distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-021-6.
CODE           Unit #31KQ06234.
MANUFACTURER   American Red Cross, Buffalo, New York.
RECALLED BY    Manufacturer, by telephone on March 2, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   New York.
QUANTITY       1 unit.
REASON         Blood product, which was labeled with an incorrect
               expiration date, was distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-034-6.
CODE           Unit #FE01853.
MANUFACTURER   Sacramento Medical Foundation Blood Center, Sacramento,
               California.
RECALLED BY    Manufacturer, by telephone on August 20, 1991, and by letter
               dated September 12, 1991.  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product, which was collected from a donor who was born
               in or emigrated from an area of high risk for the human
               immunodeficiency virus types 1 and 2 (HIV 1 and 2) was
               distributed.

_______________
PRODUCT        Plasma.  Recall #B-035-6.
CODE           Unit numbers:  17T49111021, 17T49819021, 17T5A510021,        
               17T5B131021, 17T5B414021.
MANUFACTURER   Plasma Alliance, Inc., Tulsa, Oklahoma.
RECALLED BY    Plasma Alliance, Inc., Knoxville, Tennessee, by letter dated
               October 21, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       5 units.
REASON         Blood products, which were collected from a donor at
               increased risk for infectious diseases, were distributed.

                                    -5-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        (a) Source Leukocytes.  Recall #B-619-5.
CODE           Unit numbers:  60FG00115, 60GR00245, 60FG00012, 60L35357,
               60GL00894, 60FN01786.
MANUFACTURER   American National Red Cross, Miami, Florida.
RECALLED BY    Manufacturer, by letters dated October 18, 1993, November
               18, 1993, December 20, 1993, February 25, 1994, March 3 and
               17, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Florida and Switzerland.
QUANTITY       6 units.
REASON         Blood products, which tested repeatedly reactive for the
               antibody to the human T-lymphotropic virus type I (anti-
               HTLV-1) tested repeatedly reactive for syphilis by the
               Olympus PK7100 Treponema pallidum (TP) test and were
               confirmed positive by the Fluorescent Treponemal Antibody
               Absorption assay (FTA); tested initially reactive for
               syphilis; were collected from a donor with a medical history
               of cancer; and were prepared greater than eight hours after
               phlebotomy, were distributed.

_______________
PRODUCT        Single Donor Plasma.  Recall #B-020-6.
CODE           Unit #53F89420.
MANUFACTURER   American National Red Cross, Baltimore, Maryland.
RECALLED BY    Manufacturer, by letters dated October 3 and 17, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Maryland.
QUANTITY       1 unit.
REASON         Blood products which were prepared greater than eight hours
               after phlebotomy, or were prepared from whole blood that was
               stored at room temperature greater than eight hours, were
               distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-022-6.
CODE           Unit #33FP16182.
MANUFACTURER   American Red Cross Blood Services, ΒΆ, Connecticut.
RECALLED BY    Manufacturer, by telephone on February 23, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Connecticut.
QUANTITY       1 unit.
REASON         A unit of Platelets, Pheresis, labeled with the incorrect
               expiration date, was distributed.

_______________
PRODUCT        (a) Whole Blood CPDA-1; (b) Red Blood Cells; (c) Red Blood
               Cells Leukocytes Removed; (d) Platelets, Pheresis; (e) Fresh
               Frozen Plasma.  Recall #B-024/028-6.

                                    -6-CODE           Unit numbers:  (a) Unit Nos. 7002682, 7002683 and 7002684
               (b) 4600173, 4600176 and 4600182, 6003902, 7002666, 7002667,
               8048789, 8048790, 8048894, 8048895, 8048896, 8048897,
               8048898, 8048900, 8048901, 8048908, 8048909,
               8048910, 8048911, 8048912, 8048913, 8048915,
               8960293, 8960294, 8960295, 8960296, 8960297,
               8960298, 8960299, 8960300, 8960301, 8960302,
               8960303, 8960304, 8960305, 8960306, 8960307,
               8960308, 8960309, 8960310, 8960311, 8960312,
               8960361, 8960363, 8960364, 8960365, 8960367,
               8960368, 8960369, 8960370, 8960371, 8960372,
               8960373, 8960374, 8960375, 8960376, 8960377,
               8960378, 8960379, 8960380, 8960381, 8960382,
               8960383, 8960384, 8960457, 8960458, 8960459,
               8960460, 8960461, 8960462, 8960463, 8960464,
               8960465, 8960466, 8960468, 8960469, 8960470,
               8960471, 8960472, 8960474, 8960475, 8960476,
               8960477, 8960478, 8960479, 8960480, 8960482,
               8960483, 8960484, 8960485, 8960486 and 8960487;
               (c) 8960390; (d) 5300363, 5300364, 5300365, 5300366,
               5300367, 5300368 and 5300369; (e) 8048891 and 8048892.
MANUFACTURER   West Tennessee Regional Blood Center, Inc., Jackson,
               Tennessee.
RECALLED BY    Manufacturer, by telephone on June 9, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Tennessee and Georgia.
QUANTITY       (a) 3 units; (b) 92 units; (c) 1 unit; (d) 7 units; 
               (e) 2 units.
REASON         Blood products, which were improperly tested for the
               antibody to the hepatitis C virus encoded antigen (anti-
               HCV), were distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-029-6.
CODE           Unit #42FR28305.
MANUFACTURER   American Red Cross Blood Services, Cleveland, Ohio.
RECALLED BY    Manufacturer, by telephone on May 12, 1994, and by letter
               dated June 2, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Ohio.
QUANTITY       1 unit.
REASON         A unit of Platelets, Pheresis which was incompatible by
               crossmatch, was labeled as compatible and distributed for
               transfusion.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen Plasma. 
               Recall #B-030/032-6.
CODE           Unit #5710125.
MANUFACTURER   BloodCare, Dallas, Texas.
RECALLED BY    Manufacturer, by letter dated October 7, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.

                                    -7-QUANTITY       1 unit of each component.
REASON         Blood products, which were collected from a donor who had
               undergone a cornea transplant within twelve months of
               donation, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-033-6.
CODE           Unit #3188432.
MANUFACTURER   Central Texas Regional Blood Center, Austin, Texas.
RECALLED BY    Manufacturer, by telephone on October 16, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Autologous blood product, which tested positive for syphilis
               but was not labeled with a "Biohazard" label, was
               distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-036-6.
CODE           Unit #GL02981.
MANUFACTURER   Blood Bank of the Alameda-Contra Costa Medical Association,
               Oakland, California.
RECALLED BY    Manufacturer, by telephone March 20, 1992.  Firm-initiated
               recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product, which was exposed to improper storage and
               unacceptable temperatures was distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Point of View Monitor, Model #0998-00-0105-01, used to
               monitor heart rate and blood pressure.  Recall #Z-044-6.
CODE           All serial numbers.
MANUFACTURER   Datascope Corporation, Paramus, New Jersey.
RECALLED BY    Manufacturer, by telephone and fax on Marc 31, 1995,
               followed by letter dated April 3, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       424 units were distributed.
REASON         The monitor resets itself, causing monitoring information to
               be suspended for up to several minutes until the unit powers
               up.  Upon reset, previously selected settings, including
               alarm settings would need to be re-established.

_______________
PRODUCT        Harris Precoat Plus Femoral Hip Stems:
               (a) Product #9028-10, Harris Precoat Plus Femoral Stem,
               Small (120mm Stem Length):
               (b) Product #9028-20, Harris Precoat Plus Femoral Stem,
               Medium (140mm Stem Length).  Recall #Z-046/047-6.

                                    -8-CODE           All lots produced on or before 12/31/93.
MANUFACTURER   Zimmer, Inc., Warsaw, New Jersey.
RECALLED BY    Manufacturer, by letters sent August 28 and 29, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Firm estimated that no more than 100 to 200 units remained
               on market at time of recall initiation.
REASON         Laser etching may cause the femoral stems to fracture in
               use, requiring revision surgery to replace the broken hip
               implant.

_______________
PRODUCT        SJS Partners Latex Examination Gloves, Powder Free, in boxes
               of 100.  Recall #Z-048-6.
CODE           Lots prefixed 406 and 407; Reorder #GWZS 2220M.
MANUFACTURER   Shanghai Latex Factory, Changzheng Branch, China.
RECALLED BY    SJS Supreme, Inc., Yorba Linda, California, by letter dated
               September 18, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   California, New Jersey, Illinois, Alabama, New York.
QUANTITY       1,100 cases (20 boxes per case) were distributed; firm
               estimates none remains on the market.
REASON         The gloves have the potential for spontaneous combustion,
               demonstrated by signs of heat and possibly smoldering within
               the cases.

_______________
PRODUCT        Datascope's System 95 Intra-Aortic Balloon Pump (IABP),
               Catalog #0998-00-0087XX, for use as a patient aid during
               pre-, intra- or post operative open heart surgery.  
               Recall #Z-049-6.
CODE           Serial numbers:  1000-1610 with the exception of the
               following units:  1051, 1062, 1176, 1211, 1220, 1222, 1230,
               1270, 1373, 1374, 1383, 1387, 1448, 1479, 1480, 1509, 1511,
               1522, 1529, 1555, 1559, 1561, 1562, 1563, 1565, 1658, 1569,
               1587, 1593, 1602, 1603, 1605, 1606.
MANUFACTURER   Datascope Corporation, Paramus, New Jersey.
RECALLED BY    Manufacturer, by letters dated May 2 and 31, 1995.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       439 units were distributed; firm estimated that 428 units
               remained on market at time of recall initiation.
REASON         Keypad failure causes the unit to become non-functional,
               resulting in an inability to trigger balloon inflation.

_______________
PRODUCT        Hewlett Packard Defibrillator Interface Cable, 
               Part #0012-00-0753-01.  Recall #Z-050-6.
CODE           None.
MANUFACTURER   National Cable & Molding, Los Angeles, California.
RECALLED BY    Datascope Corporation, Paramus, New Jersey, by letter May
               25, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.

                                    -9-QUANTITY       39 units.
REASON         Due to incorrect wiring, the defibrillator may not recognize
               the ECG input and, therefore, will not discharge.  It may
               also produce an intermittent condition which could cause a
               false trigger, resulting in a random discharge.

_______________
PRODUCT        AAMI Standard Defibrillator Interface Cable, 
               Part #0012-00-0731.  Recall #Z-051-6.
CODE           None.
MANUFACTURER   National Cable & Molding, Los Angeles, California.
RECALLED BY    Datascope Corporation, Paramus, New Jersey, by letter May
               24, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       296 units were distributed.
REASON         Due to a design flaw, there may be an intermittent
               connection between the connector's ground shield and the
               conductive coating in the Passport's housing.  This
               connection to ground can cause the random discharge of the
               defibrillator during synchronized cardioversion.

_______________
PRODUCT        ECG Electrodes:
               (a) Medi-Trace Offset Dx ECG Electrodes, under the Graphic
               Controls label;
               (b) Matrx R.A.M. Electrodes, under the Matrx Medical label;
               (c) Medi-Trace S'Offset ECG Electrodes, under the Graphic
               Controls label.  Recall #Z-057/059-6.
CODE           Order No.          Lot Numbers
               (a) 8101           5111, 5141, 5151
                   8103           5101, 5111, 5121, 5141, 5151, 5161
                   8105           5101, 5111, 5121, 5151, 5161
                   8125           5091, 5101, 5121, 5141, 5151
               (b) MPX 9114001    5111
               (c) 8201           5101, 5111, 5121, 5141, 5151, 5161
                   8203           5101, 5111, 5141,5151
                   8205           5101, 5111, 5121, 5141, 5151
                   8225           5092, 5101, 5102, 5111, 5121.
MANUFACTURER   Graphic Controls Corporation, Methuen, Massachusetts.
RECALLED BY    Graphic Controls Corporation,Buffalo, New York, by letter
               dated September 1, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,657,500 units were distributed.
REASON         The electrodes produce excessive noise and artifact due to a
               reworked silver strip component which was found to be
               unstable.

_______________
PRODUCT        Redman Motorized Wheelchairs:
               (a) Model 107RW, Road Warrior, Motorized Wheelchair
               (b) Model 107RS, Road Savage, Motorized Wheelchair
               (c) Model 107RC, Geronimo, Motorized Wheelchair

                                   -10-               (d) Model 107PR, Geronimo PR, Motorized Wheelchair
               (e) Model 107RU, Chief, Motorized Wheelchair
               (f) Model 107SR, Chief SR, Motorized Wheelchair
               (g) Model 07RP, Red Pack, Motorized Wheelchair.
               Recall #Z-060/066-6.
CODE           All wheelchairs having a Penny & Giles 60 AMP controller
               with a serial number prefix of 9106XXXX to 9311XXXX, or if
               the serial number has a "UG" suffix.
MANUFACTURER   Redman, Inc., Tucson, Arizona.
RECALLED BY    Manufacturer, by letter January 5, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Israel.
QUANTITY       361 wheelchairs were distributed.
REASON         The joystick circuitry can cause unintended starting and
               stopping of the wheelchair.  This defect presents a moderate
               risk of serious adverse health consequences.

_______________
PRODUCT        Venshield Heat and Moisture Exchange Filter.
               Recall #Z-068-6.
CODE           Lot numbers 2735 and 2747.
MANUFACTURER   Colorado MEDtech, Inc., Boulder, Colorado.
RECALLED BY    Manufacturer, by telephone during the week of August 21,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Arizona, California, Colorado, Florida, Illinois, Kentucky,
               Massachusetts, Missouri, Oklahoma, Pennsylvania, Texas.
QUANTITY       15,000 units.
REASON         The filter media can separate from the frame, obstructing
               the respiration gas passage if the filter is saturated with
               secretions.

_______________
PRODUCT        CPU Software Version 1.07A in the Hydrajust IV Urological
               Table:  (a) Part #400001; (b) Part #400002;
               (c) Part #400005; (d) Part #400006.  Recall #Z-070/073-6.
CODE           Serial Numbers for the tables are as follows:
               4793-0132  2194-0177  3794-0200  2494-0182  2194-0176
               0894-0160  2194-0175  1294-0166  4993-0127  2394-0179
               3094-0191  2494-0181  2194-0174  3794-0199  2594-0185
               2394-0180  3094-0192  3094-0188  3794-0198  2093-0117
               2494-0183  3094-0186.
MANUFACTURER   Liebel-Flarsheim Company, Cincinnati, Ohio.
RECALLED BY    Manufacturer, by information sheet dated September 26, 1994. 
               Firm-initiated field correction complete.
DISTRIBUTION   Nationwide, Mexico, Scotland.
QUANTITY       22 tables were subject to field correction. 
REASON         There is risk of patient or operator exposure when the X-ray
               tube is in the park position.


                                   -11-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Ag 4000 (Ag/AgC1) Resting ECG Electrodes Unitrace, Silver
               EEZ, Generic.  Recall #Z-054-6.
CODE           Lot numbers: UAG4-C 4178, 4201A, 4255A, 4170A, 5027; CDI
               LAG4-CG; CII LAG4-CG, 4178, 4201A, 4297, 5027, 5107A, 5135;
               CDI LAG4-C, 4178, 4201A, 4297.
MANUFACTURER   LecTec Corporation, Minnetonka, Minnesota.
RECALLED BY    Manufacturer, by letter dated August 10, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Arizona, California, Colorado, Connecticut, Florida,
               Louisiana, Michigan, Minnesota, Missouri, New York, Rhode
               Island, Texas, Wisconsin, France, Canada, Israel, Sweden.
QUANTITY       21,268,500 units were distributed.
REASON         Certain lots of the Tracests Ag400 Silver/Silver Chloride
               (Ag/AgCl) Resting ECG Electrodes may not satisfy the
               Association for the Advancement of Medical Instrumentation
               (AAMI) defibrillation overload recovery standard throughout
               the product's 24-month shelf life, which could lead to
               misdiagnosis by the clinician.

                                   -12-


END OF ENFORCEMENT REPORT FOR NOVEMBER 1, 1995.  BLANK PAGES MAY