FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
07/18/1995

 ENFORCEMENT REPORT FOR 07/19/95 


July 19, 1995                                                    95-29

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Pyridoxine HCl, 100 mg (Vitamin B-6), in bottles of 210, a
               dietary supplement, under the Great Earth label.  
               Recall #F-683-5.
CODE           DEC/97-55594.
MANUFACTURER   Phoenix Laboratories, Inc., Hicksville, New York.
RECALLED BY    Manufacturer, by letter sent March 24, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   California.
QUANTITY       237 bottles were distributed; firm estimates none remains on
               the market.
REASON         Niacin tablets were mislabeled as Vitamin B6 Tablets.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Dilantin Kapseals 100 mg (extended Phenytoin Sodium
               capsules, USP) in bottles of 100, an anti-epileptic.  
               Recall #D-210-5.
CODE           Lot #064N4FA EXP 10/96.
MANUFACTURER   Warner Lambert Company, Fajardo, Puerto Rico.
RECALLED BY    Parke Davis, Division of Warner Lambert Company, Morris
               Plain, New Jersey, by letters mailed between June 23 and 26,
               1995.  Firm-initiated recall ongoing.

DISTRIBUTION   Nationwide.
QUANTITY       39,510 bottles were distributed.
REASON         Product does not meet dissolution specifications.

_______________
PRODUCT        Quinidine Sulfate, 200 mg, Immediate Release Tablets, in
               bottles of 100, typically prescribed for initial loading
               dosage of Quinidine prior to maintenance therapy with
               Quinidine Gluconate.  Recall #D-211-5.
CODE           Lot #30373 EXP 3/97.
MANUFACTURER   Mutual Pharmaceutical Company, Philadelphia, Pennsylvania.
RECALLED BY    Manufacturer, by letter dated June 19, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,961 bottles wee distributed.
REASON         Product does not meet content uniformity specifications.

_______________
PRODUCT        Palgic D Tablets in plastic bottles of 100 with screw on
               plastic caps, an antihistamine/decongestant.
               Recall #D-212-5.
CODE           Lot #SH402CP EXP 8/96.
MANUFACTURER   Sage Pharmaceuticals, Inc., Shreveport, Louisiana.
RECALLED BY    Manufacturer, by letter June 26, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Louisiana.
QUANTITY       887,400 tablets were distributed.
REASON         Product does not meet time release specifications for
               pseudoephedrine, HCl ingredient.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Histenol Forte Tablet, an OTC multisymptom cold product,
               distributed in bulk containers, and repacked into individual
               2 tablet packets.  The packets are packaged into cartons
               containing 100, 180, 250, or 1,000 packets.  
               Recall #D-209-5.
CODE           Lot numbers:  P57 and P58.
MANUFACTURER   J.B. Laboratories, Holland, Michigan.
RECALLED BY    Manufacturer, by letter May 26, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY       Firm estimated that 200,000 tablets remained on market at
               time of recall initiation.
REASON         Presence of small amounts of Ephedrine HCl in the
               pseudoephedrine HCl raw material.


                                    -2-
RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  
               Recall #B-199/200-5.
CODE           Unit numbers:  1194301, 1194304, 1194307, 5215275, 7152323.
MANUFACTURER   Carter Blood Center, Fort Worth, Texas.
RECALLED BY    Manufacturer, by letters dated September 3, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas, Missouri.
QUANTITY       5 units of each component.
REASON         Corresponding blood product was contaminated with bacillus
               species

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  
               Recall #B-462/463-5.
CODE           Unit #47G90945.
MANUFACTURER   American Red Cross Blood Services, Peoria, Illinois.
RECALLED BY    Manufacturer, by letter dated March 15 and 21, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, California
QUANTITY       1 unit of each component.
REASON         Blood products for which initial reactive test results for
               antibody to human immunodeficiency virus type 1 (anti-HIV-1)
               were not repeated in duplicate, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #468-5.
CODE           Unit #09J82762.
MANUFACTURER   American Red Cross Blood Services, Omaha, Nebraska.
RECALLED BY    Manufacturer, by telephone July 11, 1994, followed by letter
               dated July 28, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Iowa.
QUANTITY       1 unit.
REASON         Blood product, corresponding to a unit of platelets
               contaminated with coagulase negative Staphylococcus, was
               distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Red Blood Cells.  Recall #B-461-5.
CODE           Unit numbers:  13FF20519, 13FF20523, 13FF20589, 13FF20331.
MANUFACTURER   American Red Cross Blood Services, Detroit, Michigan.
RECALLED BY    Manufacturer, by letters dated February 8, 1995, and March
               2, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       4 units.
REASON         Blood components, which were collected in expired blood
               collection bags, were distributed.

                                    -3-
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-464/465-5.
CODE           Unit numbers:  (a) 287539, 485734, 486644, 983458; 
               (b) 486644.
MANUFACTURER   W.E. and Lela I. Stewart Blood Center, Inc., Tyler, Texas.
RECALLED BY    Manufacturer, by telephone July 26, 1994 or October 5, 1994,
               or by letters dated July 27, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Texas.
QUANTITY       (a) 4 units; (b) 1 unit.
REASON         Blood products, which tested reactive for syphilis by the
               Rapid Plasma Reagin assay (RPR), were distributed.

_______________
PRODUCT        Platelets.  Recall #B-466-5.
CODE           Unit #575366.
MANUFACTURER   W.E. and Lela I. Stewart Blood Center, Inc., Tyler, Texas.
RECALLED BY    Manufacturer, by telephone September 16, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood product, which tested initially reactive for the
               antibody to the hepatitis B core antigen (anti-HBc), were
               distributed.

_______________
PRODUCT        Anti-Lewisa Blood Grouping Reagent.  Recall #B-467-5.
CODE           Lot numbers:  2347, 2348, 2349.
MANUFACTURER   Organon Teknika Corporation, West Chester, Pennsylvania.
RECALLED BY    Organon Teknika Corporation, Durham, North Carolina, by
               letters dated February 6 and 10, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       340 units of lot 2347, 1,390 units of lot 2348, 408 units of
               lot 2349 were distributed. 
REASON         Anti-Lewisa (anti-Lea) Blood Grouping Reagent, which was
               packaged with a product insert containing an incorrect
               procedure for use, was distributed.

RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Devices labeled as Selector Catheters, 100 cm, 5 French,
               Left Coronary Bypass, Catalog #445.0-062.  Recall #Z-759-5.
CODE           Lot numbers:  14058, 14274.
MANUFACTURER   NAMIC Caribe, Inc., Aguadilla, Puerto Rico.
RECALLED BY    NAMIC U.S.A. Corporation, Glens Falls, New York, by
               telephone April 18 & 19, 1995, followed by letter dated
               April 18, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Florida, Iowa, Illinois, Maryland, Missouri,
               Michigan Nebraska, New York, Washington State, Ireland.
QUANTITY       118 units were distributed

                                    -4-
REASON         Mislabeled.  Packages labeled as Selector Catheters, 100 cm,
               5 French, Left Coronary Bypass, contains an Amplatz Left 1.5
               catheter.

_______________
PRODUCT        Baxter Flow-Through Injectable Housing, Sterile, Model 93-
               505, when used in conjunction with a Baxter temperature
               probe, is a flow-through housing which allows direct
               temperature measurement of the thermal indicator at the
               proximal lumen infusion site.  Recall #Z-848-5.
CODE           Lot numbers:  4V07F-3960, 4W07F-0037.
MANUFACTURER   Baxter Healthcare Corporation of Puerto Rico, Anasco, Puerto
               Rico.
RECALLED BY    Baxter Healthcare Corporation, Irvine, California, by
               letters dated March 16 and 17, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Arkansas, California, Colorado, Florida, Hawaii, Idaho,
               Kansas, Massachusetts, Maryland, Minnesota, North Carolina,
               North Dakota, New York, Pennsylvania, Virginia, Wisconsin,
               and international.
QUANTITY       3,072 units were distributed.
REASON         The lots were found to have inadequate package seals which
               may compromise the sterility of the device.

_______________
PRODUCT        USCI 5F Envision Angiographic Catheters, intravascular
               diagnostic catheters used to record intracardiac pressures
               to sample blood and to introduce substances into the heart
               and vessels.  Recall #Z-857-5.
CODE           All units manufactured through 12/16/95.  Item numbers: 
               Item No.                Description
               051000                   Straight Pigtail
               051001                   145 degree Pigtail
               051002                   155 degree Pigtail
               051003                   JL 3.5
               051004                   JL 4.0
               051005                   JL 5.0
               051006                   JL 6.0
               051007                   No Torque Right
               051008                   JR 3.5
               051009                   JR 4.0
               051010                   JR 5.0
               051011                   JR 6.0
               051012                   Champ 0.5
               051013                   Champ 1.0
               051015                   Amplatz Right 1
               051016                   Amplatz Right 2
               051017                   Amplatz Left 1
               051018                   Amplatz Left 2
               051019                   Amplatz Left 3
               051020                   Multipurpose A-1
               051021                   Multipurpose A-2

                                    -5-
               051022                   Multipurpose B-1
               051023                   Multipurpose B-2
               051024                   Coronary Bypass
               051025                   Coronary Bypass
               051026                   Internal mammary
               051027                   Sones A
               051028                   Sones B
               051029                   Sones C
               051030                   Brachial 3.5
               051031                   Brachial 4.0
               051032                   Brachial 4.5
               051033                   155 degree Pigtail
               051034                   Tight Radius Straight Pigtail
               051035                   Tight Radius 145 degree Angle
               051037                   Champ 3.0
               051038                   Champ 3.5
               051043                   Modified JR 4.0
               355050                   EZ Pak w/.035"wire and Straight
                                        Pigtail
               355051                   EZ Pak w/.035" wire and 145 degree
                                        angle Pigtail
               355052                   EZ Pak w/.035" wire and 155 degree
                                        angle Pigtail
               385050                   EZ Pak w/.038" wire and Straight
                                        Pigtail
               385051                   EZ Pak w/.035" wire and 145 degree
                                        angle Pigtail
               385052                   EZ Pak w/.038" wire and 155 degree
                                        angle Pigtail
               551050                   Angio Kit with Straight Pigtail
               551051                   Angio Kit with 145 degree Angle
                                        Pigtail
               551052                   Angio Kit with 155 degree Angle
                                        Pigtail.
MANUFACTURER   USCI Manufacturing Facility, Billerica, Massachusetts.
RECALLED BY    USCI Division, C.R. Bard, Inc., Billerica, Massachusetts, by
               letter January 6, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       15,020 units were distributed.
REASON         The tip segment separates from the proximal shaft.

_______________
PRODUCT        USCI 5F Envision Angiographic Catheters, intravascular
               diagnostic catheters used to record intracardiac pressures
               to sample blood and to introduce substances into the heart
               and vessels.  Recall #Z-858-5.
CODE           All units manufactured from 12/16/94 - 5/12/95.  
               Item No.                Description
               051000                   Straight Pigtail
               051001                   145 degree Pigtail
               051002                   155 degree Pigtail
               051003                   JL 3.5

                                    -6-
               051004                   JL 4.0
               051005                   JL 5.0
               051006                   JL 6.0
               051007                   No Torque Right
               051008                   JR 3.5
               051009                   JR 4.0
               051010                   JR 5.0
               051011                   JR 6.0
               051012                   Champ 0.5
               051013                   Champ 1.0
               051015                   Amplatz Right 1
               051016                   Amplatz Right 2
               051017                   Amplatz Left 1
               051018                   Amplatz Left 2
               051019                   Amplatz Left 3
               051020                   Multipurpose A-1
               051021                   Multipurpose A-2
               051022                   Multipurpose B-1
               051023                   Multipurpose B-2
               051024                   Coronary Bypass
               051025                   Coronary Bypass
               051026                   Internal mammary
               051027                   Sones A
               051028                   Sones B
               051029                   Sones C
               051030                   Brachial 3.5
               051031                   Brachial 4.0
               051032                   Brachial 4.5
               051033                   155 degree Pigtail
               051034                   Tight Radius Straight Pigtail
               051035                   Tight Radius 145 degree Angle
               051037                   Champ 3.0
               051038                   Champ 3.5
               051043                   Modified JR 4.0
               355050                   EZ Pak w/.035"wire and Straight
                                        Pigtail
               355051                   EZ Pak w/.035" wire and 145 degree
                                        angle Pigtail
               355052                   EZ Pak w/.035" wire and 155 degree
                                        angle Pigtail
               385050                   EZ Pak w/.038" wire and Straight
                                        Pigtail
               385051                   EZ Pak w/.035" wire and 145 degree
                                        angle Pigtail
               385052                   EZ Pak w/.038" wire and 155 degree
                                        angle Pigtail
               551050                   Angio Kit with Straight Pigtail
               551051                   Angio Kit with 145 degree Angle
                                        Pigtail
               551052                   Angio Kit with 155 degree Angle
                                        Pigtail.

                                    -7-
MANUFACTURER   USCI Manufacturing Facility, Billerica, Massachusetts.
RECALLED BY    USCI Division, C.R. Bard, Inc., Billerica, Massachusetts, by
               letter May 23, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       7,140 units were distributed.
REASON         Pinhole defects were found in the pigtail catheters.

_______________
PRODUCT        Orthodontic wire distal-end mini cutter used by
               dentists/orthodontists, to cut different sizes of dental
               braces wire material inside the patient's mouth:
               (a) Model AEZ Part #803-0510, Ezcurra Distal Cutter with
               hold long handle; (b) Model AEZ Part 803-0511, Ezcurra
               Distal End Cutter with hold short handle.
               Recall #Z-868/869-5.
CODE           Batch numbers:  2xxx, 3xxx, 4Axx, 4Bxx, 4Cxx, 4Dxx, 4Exx,
               4Fxx, 4Gxx, 4Hxx (where the x represents any letter or
               number).  There are no serial numbers or expiration/use-by
               dates.
MANUFACTURER   ORMCO Corporation, Glendora, California.
RECALLED BY    Manufacturer, by letters dated May 5 and 11, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 5,700 cutters were distributed.
REASON         The device's tip could break in patient's mouth.

_______________
PRODUCT        Nuclear Imaging Systems:  (a) Model 1000S; (b) Model 2000S;
               (c) Model 1000XP; (d) Mode 2000XP.  Recall #Z-939/942-5.
CODE           All serial numbers.
MANUFACTURER   Ohio Imaging Nuclear Medicine Division, Picker
               International, Bedford Heights, Ohio.
RECALLED BY    Manufacturer, by letter February 1, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       252 units were distributed.
REASON         The camera's detector head may drift beyond acceptable
               limits which will result in a high force being placed
               against the patient's head or body.

_______________
PRODUCT        Endoscope Reprocessor (Ultrasonic Washer):  
               (a) System 83; (b) System 83 Plus.  Recall #Z-958/959-5.
CODE           All product manufactured before 1/92 when a design change
               was introduced.  All units with serial numbers in which the
               last two digits are lower than 92.
MANUFACTURER   Custom Ultrasonics, Inc., Ivyland, Pennsylvania.
RECALLED BY    Manufacturer, by memorandum dated April 7, 1994.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 1,200 units were distributed.

                                    -8-
REASON         Device failed to adequately clean the inner channels of the
               reprocessed endoscope, resulting in contaminated endoscopes.

_______________
PRODUCT        Sterile Non-Latex Surgical Gloves, indicated for use by
               surgeons as infection barriers during surgical procedures:
               (a) Tactyl 1, Catalog #41070; (b) Tactyl 2, Catalog #51070;
               (c) Tactylon, Catalog #51175.  Recall #Z-960/962-5.
CODE           Lot numbers:  (a) 5024-0005; (b) 5024-0003; (c) 5054-0002,
               5024-0002.
MANUFACTURER   Tactyl Technologies, Inc., Vista, California.
RECALLED BY    Manufacturer, by letter March 21, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Arizona, France, Sweden.
QUANTITY       292 shipping cases were distributed; firm estimated that on
               April 14, 1995 35 shipping cases remained on the market.
REASON         The gloves did not receive the full irradiation dose,
               thereby compromising the sterility of the gloves.

_______________
PRODUCT        Green Maxon Absorbable Sutures, Size 01 (sterile), indicated
               for use in all types of soft tissue approximation, but not
               in cardiovascular, ophthalmic or microsurgery and not in
               neural tissue.  Recall #Z-963-5.
CODE           Product #6315-72, Lot numbers:  878309 EXP 9/95, 878308 EXP
               10/95.
MANUFACTURER   Davis & Geck, Manati, Puerto Rico.
RECALLED BY    Davis & Geck, Danbury, Connecticut, by letter June 8, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       183 boxes (24 units per box) were distributed; firm
               estimates none remains on the market.
REASON         The sutures exhibit low out of package tensile strength.

_______________
PRODUCT        Davis & Geck LAPRO-CLIP Ligating Clip Remover (non-sterile). 
               Recall #Z-964-5.
CODE           Product #8487-99, lot numbers:  891890, 892262, 892381,
               893528.
MANUFACTURER   Mediscope Manufacturing, Inc., North Bergen, New Jersey
               (contract manufacturer).
RECALLED BY    Davis & Geck, Danbury, Connecticut, by letters sent June 2,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Georgia, Illinois, North Carolina, Louisiana,
               Massachusetts.
QUANTITY       26 units were distributed; firm estimated that 20 devices
               remained in the field at time of recall initiation.
REASON         The grasping hook of the device may break and become
               detached from the device.

                                    -9-
_______________
PRODUCT        Philips Medical Systems SL Series Multi Energy Radiotherapy
               Systems.  Recall #Z-987-5.
CODE           Serial numbers 5001 to 5280.
MANUFACTURER   Philips Medical Systems Radiotherapy, West Sussex, England.
RECALLED BY    Philips Medical Systems North America Company, Shelton,
               Connecticut, by visit after February 1994.  Firm-initiated
               field correction complete.
DISTRIBUTION   Nationwide.
QUANTITY       82 systems were distributed.
REASON         There is a potential for a 10 second delay in the "No Dose
               Interlock" mode if the ion chamber -300 volt supply fails,
               which could continue radiation for an extra 10 seconds.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Lithium Serum Flame Standard for Flame Photometer. 
               Recall #Z-965-5.
CODE           Lot #DB-1.
MANUFACTURER   Radiometer Copenhagen, Copenhagen, Denmark.
RECALLED BY    Radiometer America, Westlake, Ohio, by telephone on or about
               December 7, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Maine, Illinois, Oregon.
QUANTITY       8 bottles were distributed.
REASON         The standard has a lithium content that is lower than the
               specified nominal value.  This can cause the lithium reading
               to be higher than the nominal variation allowed.

_______________
PRODUCT        Qualicheck 3 Quality Control Material for ABL-520, control
               system designed to monitor the accuracy of pH, blood gas,
               and oximetry measurements in the analyzer.  Recall #Z-966-5.
CODE           Lot #44.
MANUFACTURER   Radiometer A/S EmdruppvJ 72, Copenhagen NV, Denmark.
RECALLED BY    Radiometer America, Westlake, Ohio, by letter October 28,
               1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       156 boxes were distributed; firm estimates none remains on
               the market.
REASON         The MetHb (Methemoglobin) value for the material was outside
               the stated control range.  The actual value was higher than
               the stated value.

_______________
PRODUCT        Qualicheck 4 Material for ABL-505, a quality control
               standard for pH, blood gases, and electrolytes.  
               Recall #Z-967-5.
CODE           Lot #27.
MANUFACTURER   Radiometer A/S Emdruppvej 72, Copenhagen NV, Denmark.
RECALLED BY    Radiometer America, Westlake, Ohio, by letter August 6,
               1993.  Firm-initiated recall complete.

                                   -10-
DISTRIBUTION   California and Texas.
QUANTITY       13 boxes were distributed; firm estimates none remains on
               the market.
REASON         The potassium value is lower than labeled.

_______________
PRODUCT        Dader Thromboplastin-IS (TPS), in-vitro diagnostic product
               for use in prothrombin time (PT) determination and
               prothrombin time based assays.  Recall #Z-968/973-5.
CODE           Catalog No. B4213-20, Lot Nos. TPS-75, TPS-80, TPS-82;
               Catalog No. B4213-50, Lot Nos. TPS-77, TPS-79,  TPS-79Z,
                                              TPS-81,TPS-86;
               Catalog No. B4213-100, Lot No. TPS-74;
               Catalog No. 281052, Lot Nos.   TPS-82, TPS-87;
               Catalog No. 281053, Lot Nos.   TPS-75, TPS-77, TPS-79,
                                              TPS-81, TPS-86;
               Catalog No. 281054, Lot No.    TPS-74.
MANUFACTURER   Dade International of Puerto Rico, Aguada, Puerto Rico.
RECALLED BY    Dade International, Inc., Miami, Florida, by letter dated
               May 1995, and by Fax to international distributors.
DISTRIBUTION   Nationwide and international.
QUANTITY       44,161 packages.
REASON         The device fails to meet the labeled stability claim after
               reconstitution when incubated.

_______________
PRODUCT        Dade Thromboplastin-IS (TPS), in-vitro diagnostic product
               for use in prothrombin time (PT) determination and
               prothrombin time based assays.  Recall #Z-974/979-5.
CODE           Catalog No. B4213-100, Lot No. TPS-64;
               Catalog No. 281054, Lot Nos. TPS-64, TPS-68, TPS-73, TPS-78;
               Catalog No. B4213-50, Lot No. TPS-67;
               Catalog No. 281053, Lot No. TPS-67;
               Catalog No. 281052, Lot No. TPS-70;
               Catalog No. B4213-20, Lot No. TPS-84.
MANUFACTURER   Dade International of Puerto Rico, Aguada, Puerto Rico.
RECALLED BY    Dade International, Inc., Miami, Florida, by letter dated
               June 1995, and by fax to international distributors.
DISTRIBUTION   Nationwide, Canada, Mexico, Costa Rica, Chile, Colombia,
               Japan.
QUANTITY       23,576 packages were distributed.
REASON         The device fails to meet the labeled stability claim after
               reconstitution when incubated.
_______________
PRODUCT        Olympic EVIS Video Endoscopes, used for endoscopic diagnosis
               and treatment within the lower digestive tract (colon and
               rectum):  (a) Model PCF-130I; (b) Model PCF-130L.
               Recall #Z-988/989-5.
CODE           Lot numbers D & G.
MANUFACTURER   Olympus Optical Company, Ltd., Tokyo, Japan.
RECALLED BY    Olympus America, Inc., Melville, New York, by letter May 24,
               1995.  Firm-initiated recall ongoing.
                                   -11-
DISTRIBUTION   Nationwide and Puerto Rico.
QUANTITY       37 were distributed.
REASON         The video endoscopes are more susceptible to surges of
               electrical current due to a manufacturing defect in the
               charged coupled device (CCD) chip which will cause the video
               endoscopes to lose their images.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
===========
_______________
PRODUCT        Vitamin K Injectable, 100 ml.  Recall #V-034-5.
CODE           Lot numbers:  3400, 3480, 3558, 3584.
MANUFACTURER   GEN/Rx (formerly AVP), Fort Collins, Colorado.
RECALLED BY    manufacturer, by letter mailed June 15, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   California, Colorado, Florida, Iowa, Kansas, Missouri,
               Nebraska, New Jersey, New York, Ohio, Oklahoma, Texas, South
               Carolina, Puerto Rico.
QUANTITY       5,625 units of lot 3400; 3,459 units of lot 3480; 3,601
               units of lot 3,558; 3,036 units of lot 3584 were
               distributed.
REASON         Product failed stability testing, falling below
               specifications.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS III
==========
_______________
PRODUCT        Coop Lamb Creep Feed Medicated packaged in 50 pound bags. 
               Recall #V-033-5.
CODE           G12234 and L17644.
MANUFACTURER   Farmland Industries, Inc., Kansas City, Kansas.
RECALLED BY    Manufacturer, by telephone January 30-31, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Oklahoma, Missouri, Kansas.
QUANTITY       328 bags were distributed.
REASON         Superpotent.

                                   -12-


END OF ENFORCEMENT REPORT FOR JULY 19, 1995.  BLANK PAGES MAY FOLLOW.

                                   ####