FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
05/31/1995

ENFORCEMENT REPORT FOR 05 31 95

May 31, 1995                                               95-22

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Bi-Lo Orchard Harvest Tomato Juice, in 46 ounce cans. 
               Recall #F-665-5.
CODE           Lot #94092OR1.
MANUFACTURER   Redwing Company, Geneva, Indiana.
RECALLED BY    Bi-Lo, Inc., Mauldin, South Carolina, by computer mail
               system on January 25, 1995, followed by hard copy of letter. 
               Firm-initiated recall complete.
DISTRIBUTION   North Carolina, South Carolina, Georgia.
QUANTITY       582 cases were distributed.
REASON         The label declares sodium content as zero "0" when it
               actually contains 760 mg of sodium.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        G.R. Diamond New York Icicle, 10 Assorted Freeze Pops,
               artificial flavors, net weight 28 fluid ounces, and 24 fluid
               ounces in poly bags.  Recall #F-659-5.
CODE           None.
MANUFACTURER   Great Rich Trading Company, Inc., Thailand.
RECALLED BY    Great Rich Trading Company, Inc., Brooklyn, New York, by
               letter January 12, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Undetermined.
QUANTITY       Undetermined.
REASON         Product contained undeclared colors.

_______________
PRODUCT        Ultra Slim-Fast Fat Free Breakfast Bars, sold in packages of
               six in the following flavors:  (a) Apple Fruit Bars; 
               (b) Fig Fruit Bars; (c) Blueberry Fruit Bars; (d) Strawberry
               Fruit Bars.  Recall #F-660/663-5.
CODE           (a) FFSA 4326 through 5070; (b) FFSF 4326 through 5070;
               (c) FFSB 4326 through 5070; (d) FFSS 4326 through 5070.
MANUFACTURER   Fleetwood Snacks, Inc., Fleetwood, Pennsylvania.
RECALLED BY    Slim-Fast Foods Company, West Palm Beach, Florida, by
               telephone beginning on April 25, 1995, followed by letter. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       10,596,168 breakfast bars were produced; firm estimated that
               3,000,000 to 5,000,000 bars remained on market at time of
               recall initiation.
REASON         Product is contaminated with plastic pieces.

_______________
PRODUCT        Bumble Bee brand Solid White Tuna in water, in 6-1/8 ounce
               cans.  Recall #F-664-5.
CODE           Lot numbers:  SWP3J 67HD1 on can and 67HD1 23021 on case.
MANUFACTURER   Bumble Bee International, Inc., Mayaquez, Puerto Rico.
RECALLED BY    Winn Dixie Distribution Center, Hammond, Louisiana, by
               visit. Firm-initiated recall complete.
DISTRIBUTION   Louisiana, Mississippi.
QUANTITY       480 cases (per cans per case) were distributed; firm
               estimates none remains on the market.
REASON         Product was adulterated due to a petroleum distillate odor
               which renders it unfit for human consumption.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Recovered Plasm.  Recall #B-336-5.
CODE           C13094, C13196, C13228, C13283, C13669.
MANUFACTURER   Sewickley Valley Hospital Blood Bank, Sewickley,
               Pennsylvania.
RECALLED BY    Manufacturer, by letter February 28, 1995, and by fax March
               1, 1995 complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       5 units.
REASON         Blood products, which were collected from autologous donors
               who did not meet the criteria for homologous donation, were
               distributed.

_______________
PRODUCT        Platelets.  Recall #B-348-5.
CODE           Unit #R71508.
MANUFACTURER   Heartland Blood Center, Aurora, Illinois.

                                    -2-
RECALLED BY    Manufacturer, by telephone February 14, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit.
REASON         Blood product, which tested negative for antibody to human
               immunodeficiency virus type 1 (anti-HIV-1), but was
               collected from a donor who previously tested repeatedly
               reactive for anti-HIV-1, was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Recovered Plasma.  Recall #B-337-5.
CODE           C13110, C13199, C13212, C13299, C13395, C13397, C13399,
               C13473, C13474, C13506, C13576, C13584, C13620, C14284.
MANUFACTURER   Sewickley Valley Hospital Blood Bank, Sewickley,
               Pennsylvania.
RECALLED BY    Manufacturer, by letter February 28, 1995, and by fax March
               1, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Pennsylvania.
QUANTITY       14 units.
REASON         Blood products, collected from autologous donors who have a
               medical history of cancer, were distributed.

_______________
PRODUCT        Spectrogen-Duo Reagent Red Blood Cells.  Recall #B-338-5.
CODE           Lot #95005 and G95005.
MANUFACTURER   Organon Teknika Corporation, West Chester, Pennsylvania.
RECALLED BY    Organon Teknika Corporation, Durham, North Carolina, by
               facsimile on December 29, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       619 units of lot 95005 and 74 units of lot G95005.
REASON         Reagent Red Blood Cells, which were found to be hemolyzed,
               were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
               Recall #B-339/341-5.
CODE           Unit #29116-6626. 
MANUFACTURER   United Blood Services Blood Systems, Inc., Chicago,
               Illinois.
RECALLED BY    Manufacturer, by letters of January 13 and 20, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, Switzerland.
QUANTITY       1 unit of each component.
REASON         Blood products, which were improperly tested for syphilis,
               were distributed.

                                    -3-
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen Plasma;
               (d) Recovered Plasma.  Recall #B-342/345-5.
CODE           Unit numbers:  (a) 1043764, 1064205, 1101529, 1116407,
               1140764, 4059997, 2092488, 1137624, 1159943, 1165709,
               2053666, 1115610, 1137619, 1144439, 1151296, 1159997;
               (b) 1116407, 1140764, 2053666, 1115610, 1137619, 1144439,
               1151296, 1159997; (c) 1140764, 1165709; (d) 1043764,
               1064205, 1101529, 1116407, 4059997, 2092488, 1137624,
               1159943, 2053666, 1115610,1137619, 1144439, 1151296,
               1159997.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City,
               Missouri.
RECALLED BY    Manufacturer, by letters of January 16, 1995, February 16,
               1995, May 12 and 13, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Kansas, Missouri, Tennessee. 
QUANTITY       (a) 16 units; (b) 8 units; (c) 2 units; (d) 14 units.
REASON         Blood products, which tested negative for hepatitis B
               surface antigen (HBsAg) and antibody to hepatitis B core
               antigen (anti-HBc), but were collected from donors who
               previously tested repeatedly reactive for HBsAg and anti-
               HBc, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-346/347-5.
CODE           Unit #3249668.
MANUFACTURER   Central Indiana Regional Blood Center, Inc., Indianapolis,
               Indiana.
RECALLED BY    Manufacturer, by telephone December 23, 1992.  Firm-
               initiated recall complete.
DISTRIBUTION   Indiana.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from a donor who had been
               immunized one week prior to donation with Rubella vaccine,
               were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Groshong Central Venous Catheters (8 Fr.) and Port Kits;
               (a) Model No. 60283 Davol MRI Subcutaneous Port with
               Attachable 1.5 mm I.D. Groshong Venous Catheter and 8 Fr.
               Introducer Kit; 
               (b) Model No. 60312 Davol Specialty Access Products CV
               Catheter 8 Fr. with Groshong Valve Long Term, Single Lumen
               Intro-eze Percutaneous Introducer System; 
               (c) Model No. 60316 Davol Dome Implanted Port with
               Attachable 1.5 I.D. Groshong Venous Catheter and Intro-eze
               Percutaneous Introducer System; 

                                    -4-
               (d) Model No. 60453 Davol MRI Hard Base Implanted Port with
               Attachable 8 Fr. Groshong Single Lumen Venous Catheter and
               Intro-Eze Percutaneous Introducer System;
               (e) Model No. 1180 Cath-tech CV Catheter with Groshong Valve
               Long Term Single Lumen Percutaneous Placement Procedure
               Tray;
               (f) Model No. 7711802 Davol CV Catheter 8 Fr. with Groshong
               Valve and VitaCuff Antimicrobial Cuff Long Term Single Lumen
               Percutaneous Placement Procedure Tray.  Recall #Z-721/726-5.
CODE           Lot numbers:  (a) 36JE2705; (b) 36KE2888; (c) 36JE2702; 
               (d) 36KE3048; (e) 36JE2689; (f) 36JE2426.
MANUFACTURER   Bard Access System, Inc., Salt Lake City, Utah.
RECALLED BY    Manufacturer, by telephone and by letter starting December
               20, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Ireland.
QUANTITY       768 units were distributed.
REASON         Particulates were found in the cannulae.

_______________
PRODUCT        Collimator Cart (a) for all ADAC Argus and (b) some ADAC
               Cirrus Nuclear Imaging Systems.  The cart is used to
               transport the collimator (lead shield) between imaging
               applications.  Recall #Z-753/754-5.
CODE           (a) Model ADAC Part #2145-5453; 
               (b) Model ADAC Part #2145-5454.
MANUFACTURER   ADAC Laboratories, Milpitas, California.
RECALLED BY    Manufacturer, by memorandum March 3, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 88 carts; (b) 101 carts were distributed.
REASON         The wheel attached to the base of the collimator transport
               cart for the Argus and Cirrus nuclear imaging systems may
               become dislodged, which may result in the collimator tipping
               over and possibly causing injury to the operator of the
               equipment.

_______________
PRODUCT        Radiation Therapy Planning Systems (RTP):  (a) RTP-5000; (b)
               XL-PLAN-73; (c) CAD PLAN, used to calculate radiation
               dosage.  These systems are NOT used with any of ADAC's
               nuclear imaging systems.  Recall #Z-755/757-5.
CODE           Software versions 3C.1, all revisions A through E.
MANUFACTURER   ADAC Laboratories, Milpitas, California.
RECALLED BY    Manufacturer, by sending Product Warning Bulletin dated May
               19, 1994.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       128 systems were distributed.


                                    -5-
REASON         Under certain conditions a software calculation error could
               occur, causing a possible bracytherapy plan which has the
               incorrect number of sources, and therefore, the incorrect
               dose rate and dose distribution.

_______________
PRODUCT        Landmark Midline Catheter, a venous access device designed
               for peripheral infusion of general IV therapy solutions and
               blood sampling.  Recall #Z-758-5.
CODE           Lot numbers earlier/lower than 404011 or 404011M.
MANUFACTURER   Menlo Care, Inc., Menlo Park, California.
RECALLED BY    Manufacturer, by letter sent between March 6 and 10, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Firm estimated that 4,400 catheters to be in consignees'
               inventory on or about March 20, 1995.
REASON         Labeling is inadequate for intended use.

_______________
PRODUCT        (a) Dual Head Genesys Nuclear Medicine Imaging Unit; (b)
               Vertex Nuclear Medicine Imaging Unit.  The item under recall
               is the software which controls the detector heads (also
               known as a gantry).  Recall #Z-765/766-5.
CODE           (a) Units with software version 2.55;
               (b) Units with software version 2.16.
MANUFACTURER   ADAC Laboratories, Milpitas, California.
RECALLED BY    Manufacturer, by performing upgrades to software.  Firm-
               initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 205 units; (b) 90 units were distributed.
REASON         An error in the software could cause an uncommanded movement
               of the collimator and detectors prior to activating the
               imaging units, which could result in potential injury to
               patients.

_______________
PRODUCT        Protective Plus I.V. Catheter Safety System.  
               Recall #Z-767-5.
CODE           Product #3060.  All individual package codes with letter G
               as 5th character and shelf pack/shipping container lot codes
               from L1670 up to and including L2560.
MANUFACTURER   Critikon, Inc., Southington, Connecticut.
RECALLED BY    Critikon, Inc., Tampa, Florida, by letter December 22, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 1 million systems were distributed.
REASON         Due to a manufacturing condition, the catheter tubing may
               separate from the catheter hub.

                                    -6-
PRODUCT        Huber Needles Butterfly Sets, used for infusing fluids or
               withdrawing blood:
               (a) Catalog #IP-1001 Huber Needle Butterfly Set;
               (b) Catalog #IP-2001 Huber Needle Butterfly Set;
               (c) Catalog #IP-3001 Huber Needle Butterfly Set;
               (d) Catalog #IP-4001 Huber Needle Butterfly Set;
               (e) Catalog #IP-2000 Huber Needle Butterfly Set;
               (f) Catalog #IP-3000 Huber Needle Butterfly Set;
               (g) Catalog #IP-4000 Huber Needle Butterfly Set.
               Recall #Z-781/787-5.
CODE           Huber Needle Infusion Sets involved have Harmac lot numbers
               7125-01 and 7239-01.  The Ideas for Medicine kit lot numbers
               that contain the recalled component made by Harmac
               Industries as follows:  (a) D091594-11, D111794-11, D111894-
               11, D120194-11, D111594-11, D090294-11, D111694-11, D111894-
               31;  (b) D092394-21, D092494-21, D110194-21, D091494-21,
               D101294-21, D092194-21, D092394-21, D100494-21, D101094-21,
               D090894-21, D091094-21, D100394-21, D102794-21, D121394-21,
               D122194-21, D091294-21; (c) T110494-31, T111894-31, T120894-
               31; (d) T091694-41, T101094-41, T100594-41, T110994-41, 
               T091294-41, T110294-41, T120594-41, T092294-41, T110994-41;
               (e) D090794-20R; (f) T102494-30; (g) T092794-40, T111194-40.
MANUFACTURER   Harmac Medical Products, Inc., Buffalo, New York.
RECALLED BY    Ideas for Medicine, Clearwater, Florida, by letter April 5,
               1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,346 kits were distributed.
REASON         Incidents have been reported of a possible functional
               problem due to needle occlusion caused by a manufacturing
               defect.

_______________
PRODUCT        Uroview X-ray Imaging Systems with fixed tables:  
               (a) Model 2000; (b) Model 2500.  Recall #Z-788/789-5.
CODE           Various serial numbers.
MANUFACTURER   OEC Medical Systems, Inc., Salt Lake City, Utah.
RECALLED BY    Manufacturer, by letter dated November 15, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       494 units were distributed.
REASON         In certain situations when the table is vertical, a pin
               could re-position, causing the head to swivel down
               uncontrolled.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Centry System 3 Dialysis Control Units manufactured between
               03/23/95 and 04/06/95:
               (a) Part #333103001; (b) Part #333100011;
               (c) Part #333100012; (d) Part #333100013;

                                    -7-
               (e) Part #333100017; (f) Part #333100101;
               (g) Part #333103111; (h) Part #333104111;
               (i) Part #333100111.  Recall #Z-768/776-5.
CODE           Nine part numbers involving 136 individually serial numbered
               units.
MANUFACTURER   Cobe Renal Care, Inc., Lakewood, Colorado.
RECALLED BY    Manufacturer, by memorandum April 20, 1995.  Firm-initiated
               field correction ongoing.
DISTRIBUTION   California, North Carolina, Indiana, Illinois, Georgia,
               Thailand, Hong Kong, Israel, Greece.
QUANTITY       34 units were distributed.
REASON         The incorrect operator warning label was placed on the cover
               of the balance chamber subassembly.

_______________
PRODUCT        AOA brand Instant Cold Compress, single use only,
               disposable.  Recall #Z-794-5.
CODE           Catalog #4299.1AOA; Lot #950316B.
MANUFACTURER   Consolidated Products & Services, Inc., Braintree,
               Massachusetts.
RECALLED BY    Manufacturer, by letter May 5, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Florida, Georgia, Indiana, Kentucky, Maryland, Ohio,
               Oklahoma, Pennsylvania, Texas.
QUANTITY       31 cases (24 units per case) were distributed.
REASON         The product is labeled as an instant cold pack, however, the
               chemical in the pack was calcium chloride which is the raw
               material used to make chemical hot packs.  The cold packs
               turned hot instead of cold.

                                    -8-
                                   ####

END OF ENFORCEMENT REPORT FOR MAY 31, 1995.  BLANK PAGES MAY FOLLOW.