Recalls and Field Corrections:  Foods -- Class I -- 04/19/1995

April 19, 1995                                                    95-16

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        (a) French Vanilla Ice Cream; (b) Strawberry Cheesecake Ice
               Cream packaged in 3 gallon cardboard bulk containers.  
               Recall #F-608/609-5.
CODE           All bulk product manufactured in 1994 and in 1995 until
               January 27, 1995.
MANUFACTURER   Hagan Ice Cream Company, Uniontown, Pennsylvania.
RECALLED BY    Manufacturer, by issuing a "Products Information Notice" on
               February 1, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Pennsylvania, Maryland, West Virginia, Ohio, Kentucky,
               Virginia.
QUANTITY       Undetermined.
REASON         Products contain undeclared eggs.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Caviar (Sevruga fish and Beluga fish) packed in blue colored
               metal containers, net weight 2.2 kg.  Recall #F-607-5.
CODE           All coded and uncoded product imported since 9-30-94.
MANUFACTURER   Geros, Moscow, Russia.
RECALLED BY    Caspian Star Caviar, Inc., Brooklyn, New York, by telephone
               January 4, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida, New York.
QUANTITY       1,932 pounds of Sevruga Caviar and 430 pounds of Beluga
               Caviar were distributed.
REASON         Product contains borates, a nonpermitted food additive.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Sam's Choice Cola in 2 liter bottles.  Recall #F-610-5.
CODE           All lots in retail stores and warehouses.
MANUFACTURER   Cott Beverages USA, Inc., Columbus, Georgia.
RECALLED BY    Manufacturer, by telephone December 13, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and Puerto Rico.
QUANTITY       1.3 million cases (8 bottles per case) were distributed.
REASON         Product is mislabeled.  The ingredient statement on the cola
               label is actually the ingredient statement for the firm's
               orange drink.

_______________
UPDATE         Recall #F-593/595-5, Various tomatoes packed in olive oil
               which appeared in the April 5, 1995 Enforcement Report
               should read:  (b) Sun Dried Tomato Pesto in Olive Oil, 8.5
               ounces under the Trader Joe and Bella Sun Luci labels.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Dilantin (extended phenytoin sodium capsules, USP),
               Kapseals, 100 mg, in bottles of 100, used in the treatment
               of seizures.  Recall #D-112-5.
CODE           Lot #05634FA EXP 2/96.
MANUFACTURER   Warner Lambert Company, Fajardo, Puerto Rico.
RECALLED BY    Parke Davis, Division of Warner Lambert Company, Morris
               Plain, New Jersey, by letter January 26, 1995.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       40,242 bottles were distributed.
REASON         Product does not meet dissolution specifications through
               expiration date.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Robichem AC formulated products containing 100 mg
               Guaifenesin and 10 mg codeine Phosphate and 3.5 percent
               alcohol per each 5 ml, in 4 ounce, pint, and gallon
               containers.  Products are antitussive expectorants, under
               the LuChem Pharmaceuticals, and Goldline labels.
               Recall #D-113-5.
CODE           Lot numbers:  7132, 7594, 7595, 7596, 7705, 7706, 7925,
               7947, 7948, 7987, 8080, 8261.
MANUFACTURER   H.N. Norton Company, Shreveport, Louisiana.

                                    -2-RECALLED BY    Manufacturer, by letter March 1, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Alabama, Arkansas, Arizona, California, Florida, Georgia,
               Illinois, Louisiana Maryland, Michigan, Mississippi,
               Oklahoma, Texas, Virginia, Washington State.
QUANTITY       202,403 4-ounce bottles; 68,963 pint bottles, and 3,617
               gallon containers were distributed.
REASON         Product does not meet specifications through expiration
               date.

_______________
PRODUCT        Congess SR, packaged in 100 capsule bottles, Rx
               decongestant.  Recall #D-114-5.
CODE           4011352 EXP 7/1/96.
MANUFACTURER   Fleming and Company, Fenton, Missouri.
RECALLED BY    Manufacturer, by letter dated March 13, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       705 bottles were distributed.
REASON         Stability data does not support expiration date.

_______________
PRODUCT        Carbamide Peroxide Otic Drops 15 ml, packed in glass
               bottles, under the Clay Park label and the following private
               labels:  Longs Ear Drops 1/2 fluid ounce (15 ml),
               distributed by Longs Drug Stores Inc.; Valumed Ear Drops 1/2
               fluid ounce (14.79 ml) Distributed by H.L. Moore, Drug
               Exchange; Treasury Ear Drops 1/2 fluid ounce (15 ml),
               distributed by Thrift Drug Inc.; Thrifty Ear Drops 0.5 fluid
               ounce (15 ml),  distributed by Thrifty Drug Stores a
               Division of Thrifty Corporation; Select Brand Ear Wax
               Removal System Net. Cont. 0.5 fluid ounce (14.8 ml),
               distributed by Select Brand Distributors; Fays Quality Ear
               Drops 1/2 fluid ounce (14.79 ml), distributed by Fay's Drug
               Stores, Div. of Fay's Inc.;  CVS Ear Wax Removal System 0.5
               fluid ounce (14.8 ml), distributed by Consumer Value Stores,
               Division of Melville Corp.; Rite Aid Ear Drops 1/2 fluid
               ounce (15 ml), distributed by Rite Aid Corporation; The
               Medicine Shoppe Ear Drops Earwax Removal Aid Net. Cont. 1/2
               fluid ounce (14.79 ml), Distributed by Medicine Shoppe
               Int'l; PayLess Ear Drops Carbamide Peroxide Net Cont. 1/2
               ounce (15 ml), distributed by Payless Drugs Stores, Inc.;
               Goldline Ear Drops 1/2 fluid ounce (15 ml), distributed by
               Goldline Laboratories; Arbor Ear Drops 0.5 fluid ounce (15
               ml), distributed by Arbor Drugs, Inc.; REVCO Ear Drops Net.
               Wt 1/2 ounce (15 ml), distributed by Revco D.S. Inc.; Valu
               Rite Ear Drops 1/2 fluid ounce (15 ml), distributed by
               McKesson Corp.; Medalist Ear Drops 1/2 fluid ounce (15 ml),
               distributed by McKesson Corp.; K&B Ear Wax Removal System,
               Net Cont. 0.5 fluid ounce (14.8 ml). distributed by Super
               Distributors, Inc., Osco Ear Drops 1/2 fluid ounce (14.8
               ml), distributed by Osco Drug.  Recall #D-115-5. 

                                    -3-CODE           Lot #   EXP        Lot #   EXP        Lot #   EXP 
               S427    4/95       T879    9/95       V754    6/96 
               S691    5/95       T882    9/95       V931    7/96 
               S913    6/95       U302    11/95      V937    7/96 
               T106    7/95       U429    12/95      W126    8/96 
               T438    8/95       U895    2/96       W237    8/96 
               T465    8/95       V098    4/96       W366    9/96 
               T751    9/95       V582    5/96       W730    10/96 
               T843    9/95       V586    5/96       W944    11/96.
MANUFACTURER   Clay Park Laboratories, Inc., Bronx, New York.
RECALLED BY    Manufacturer, by letter April 10, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Guyana, Virgin Islands.
QUANTITY       Approximately 961,882 bottles were distributed.
REASON         Subpotency.

_______________
PRODUCT        Loxitane (Loxapine Succinate) Capsules 10 mg base, in
               bottles of 100, 1000 and unit dose 10 strips, Rx
               tranquilizer.  Recall #D-116-5.
CODE           Control numbers:  372-366 EXP 12/98 (100s),
               372-367 EXP 12/98 (1000); 372-368 EXP 12/96 (unit dose
               strips).
MANUFACTURER   Lederle Laboratories, Inc., Division of American Home
               Products, Pearl River, New York.
RECALLED BY    Manufacturer, by letter March 31, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       14,715 packages were distributed.
REASON         Marginally subpotent.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Platelet, Pheresis.  Recall #B-220-5.
CODE           Unit #12T80373.
MANUFACTURER   American Red Cross, Charlotte, North Carolina.
RECALLED BY    Manufacturer, by telephone followed by letter dated October
               4, 1994.  Firm-initiated recall complete.
DISTRIBUTION   North Carolina.
QUANTITY       1 unit.
REASON         Platelets, Pheresis, which was labeled with an incorrect
               expiration date, was distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
               Recall #B-223/225-5.
CODE           Unit #40GT58385. 
MANUFACTURER   American Red Cross Blood Services, Peoria, Illinois.
RECALLED BY    Manufacturer, by letters of December 14 and 16, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, California.
QUANTITY       1 unit of each component.
                                    -4-REASON         Blood products, which tested initially reactive for antibody
               to human immunodeficiency virus type 1 (anti-HIV-1) and were
               not retested in duplicate, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-230-5.
CODE           Unit #59G12938.
MANUFACTURER   American Red Cross Blood Services, Petoskey, Michigan.
RECALLED BY    American Red Cross Blood Services, Flint, Michigan, by
               letter September 28, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit.
REASON         Blood product, collected from a donor who had taken
               antimalarial prophylaxis and traveled to an area designated
               as endemic for malaria, was distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma, Liquid; (c) Whole
               Blood.  Recall #B-231/233-5.
CODE           Unit numbers:  (a) A10471; (b) A10471; (c) M56469.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City,
               Missouri.
RECALLED BY    Manufacturer, (a&b) by letters dated May 12 and 13, 1994;
               (c) by letter dated January 16, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Arizona, Missouri.
QUANTITY       1 unit of each component.
REASON         Blood products, collected from donors with a history of
               hepatitis, were distributed for transfusion and for further
               manufacturing into injectable or non-injectable products.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Red Blood Cells.  Recall #B-219-5.
CODE           Unit #KH09226.
MANUFACTURER   Central California Blood Center, Fresno, California.
RECALLED BY    Manufacturer, by letter July 27, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product which was exposed to unacceptable storage
               temperatures was distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-221-5.
CODE           Unit numbers:  7524780, 7524796.
MANUFACTURER   Community Blood Center of Greater Kansas City, Kansas City
               Missouri.
RECALLED BY    Manufacturer, by telephone February 14, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Missouri.
QUANTITY       2 units.
                                    -5-REASON         Blood products, stored under unmonitored temperature
               conditions, were distributed.

_______________
PRODUCT        Blood Bank Quality Assurance (BBQA II) Testing Reagents. 
               Recall #B-222-5.
CODE           Lot numbers:  BBK-532, BBK-533.  Both expired 1/21/95.
MANUFACTURER   Baxter Diagnostics, Inc., Miami, Florida.
RECALLED BY    Manufacturer, by letter mailed on or about December 15,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       385 kits were distributed; firm estimates none remains on
               market.
REASON         Blood Bank Assurance II Testing Regents which were found to
               be hemolyzed were distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-226-5.
CODE           Unit #9308896.
MANUFACTURER   Hoxworth Blood Center, University of Cincinnati Medical
               Center, Cincinnati, Ohio.
RECALLED BY    Manufacturer, by letter March 7, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Florida.
QUANTITY       1 unit.
REASON         Blood product, collected from a donor with a reported
               history of cancer, was distributed.

_______________
PRODUCT        Platelets, Pheresis.  Recall #B-227-5.
CODE           Unit #53P46623.
MANUFACTURER   American Red Cross Blood Services, Baltimore, Maryland.
RECALLED BY    Manufacturer, by letter dated December 12, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Washington, D.C.
QUANTITY       1 unit.
REASON         Unlicensed blood product was distributed interstate.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.  
               Recall #B-228/229-5.
CODE           Unit #47G91031.
MANUFACTURER   American Red Cross Blood Services, Galesburg, Illinois.
RECALLED BY    American Red Cross Blood Services, Peoria, Illinois, (a) by
               telephone December 6, 1994, followed by letter December 7,
               1994; (b) by telephone December 6, 1994, followed by letter
               December 9, 1994.  Firm-initiated recall complete.
DISTRIBUTION   (a) Iowa; (b) California.
QUANTITY       1 unit of each component.
REASON         Blood components which tested initially reactive for
               hepatitis B surface antigen (HBsAg), and were not retested
               as required, were distributed.

                                    -6-RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Renasol Bicarbonate Concentrate, Product #BC-1-R, in drums
               labeled as containing 11.60 Kg of sodium chloride and 32.45
               kg of sodium bicarbonate.  Recall #Z-557-5.
CODE           Lot #06Y407.
MANUFACTURER   Renal Systems, Division of Minntech Corporation, Plymouth,
               Minnesota.
RECALLED BY    Manufacturer, by telephone December 30, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Indiana, New York, Oklahoma, Rhode Island.
QUANTITY       90 drums.
REASON         A tare weight was not used during the filling process
               shorting the drum by 2.8 Kg. of sodium chloride.  This would
               result in a sodium level 2% lower and a chloride level 4.3%
               lower in dialysate solutions prepared from this concentrate.

_______________
PRODUCT        Borning and Hill-Rom "Stabilet" Radiant Infant Warmers, used
               to achieve/maintain adequate neonate body temperatures:
               (a) Borning Stabilet, Model 200; (b) Borning Stabilet, Model
               300; (c) Hill-Rom Stabilet, Model 2000/3000;
               (d) Hill-Rom Stabilet, Model 2200/3200;
               (e) Hill-Rom Stabilet, Model 1500;
               (f) Hill-Rom Stabilet, Model 1250.  Recall #Z-594/599-5.
               (Model numbers may be followed by A, B, or C.)
CODE           All heater elements with the date code of 0294 which
               indicates the date of manufacture.
MANUFACTURER   Baume Corporation, Grasonville, Maryland (heater element).
RECALLED BY    Hill-Rom Company Inc., Batesville, Indiana, by sending
               Technical Bulletin, dated June 9, 1994.  Firm-initiated
               field correction complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       341 heater elements were distributed.
REASON         Pieces of hot quartz glass can fall from the device's heater
               element.

_______________
PRODUCT        Compass Model VM 200 Patient Gas Volume Monitor, designed
               for use with the Newport "Wave" Mechanical Ventilator.
               Recall #Z-637-5.
CODE           Serial numbers:  9403VM063, 9403VM067, 9403VM068, 9403VM069,
               9403VM070, 9403VM085, 9403VM056, 9404VM101, 9404VM102,
               9404VM103, 9404VM104.
MANUFACTURER   Newport Medical Instruments, Inc. (NMI), Costa Mesa,
               California.
RECALLED BY    Manufacturer, by letters of June 27, 1994, and August 1,
               1994.  Firm-initiated field correction complete.
DISTRIBUTION   Texas, California, Florida, South Africa.
QUANTITY       10 units were distributed.  No unrepaired units remain in
               commerce.
REASON         Product contains defective software.
                                    -7-_______________
PRODUCT        Olympus Heat Probe Unit (HPU) Instruction Manuals.  Heat
               probe unit is used as a cauterizer during surgical
               procedures in the gastrointestinal tract.  Recall #Z-643-5.
CODE           Catalog #33100.   All units.
MANUFACTURER   Olympus America, Inc., Rio Rancho, New Mexico;
               Shirakawa Olympus Co., Ltd., Japan.
RECALLED BY    Olympus America, Inc., Lake Success, New York, by letter
               sent on or about December 30, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide, Canada, South America.
QUANTITY       3,160 units were distributed nationwide.  Total units
               distributed internationally is undetermined.
REASON         The internal water lines may be contaminated with
               microorganisms.

_______________
PRODUCT        Software Versions 2000-9A, 2000-10, and 2000-11 for the
               Model 2000 Controller Microprocessor used with the Model
               1000 series and Model 2000 series Reverse Osmosis Water
               Purification Systems for dialysis applications.
               Recall #Z-644-5.
CODE           Software distributed with 133 controllers within the serial
               number range of 5026 through 5232.
MANUFACTURER   Continental Water Systems, San Antonio, Texas.
RECALLED BY    Manufacturer, by letters of January 24, 1995, and February
               20, 1995.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide, Australia, Mexico.
QUANTITY       133 units.
REASON         Software revisions were made but never validated, which
               could possibly lead to patient exposure to heavy metals and
               bacteria.

_______________
PRODUCT        Bacteria Filter with Green Tube for Expiratory Pressure
               Transducer, a sub-assembly used with the Servo Ventilator
               300 device.  Recall #Z-645-5.
CODE           Order #6079979.  Batch numbers:  60P100, 08P101, 09P102,
               10P101, 11P102, 12P101, 02R101, 02R103TP31/V, 03R102TP31/V,
               05R100TP31/V.
MANUFACTURER   Siemens-Elema AB, Life Support Systems, Division Solna,
               Sweden.
RECALLED BY    Siemens Medical Systems (SMS), Inc., Danvers, Massachusetts,
               by letter February 10, 1994, July 6, 1994, and November 29,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide. 
QUANTITY       Undetermined.
REASON         The tube connector, which allows the green tube to be placed
               in the expiratory channel of the Servo Ventilator 300, has
               been determined to have been improperly manufactured and can
               potentially flex and occlude when subjected to heat and
               humidity, which is present in the exhaled gases of
               ventilator patients.
                                    -8-_______________
PRODUCT        Components of the Sensormedics Model 3100A HFO Neonatal
               Ventilator: (a) Patient Circuit Body Part #766895, (b)
               Bellows/Watertrap, Part #766897.  Recall #Z-646-5.
CODE           Lot numbers:  (a) 110794, 111094, 111494; (b) 110294,
               111194.
MANUFACTURER   Contech Plastics, Yorba Linda, California (elbow fitting).
RECALLED BY    Sensormedics Corporation, Yorba Linda, by letter dated
               January 13, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       99 boxes of patient circuit body and 84 boxes of the
               Bellows/Watertrap were distributed; FDA estimates little
               product remains on the market.
REASON         The elbow fitting for the patient circuit and bellows
               assembly is totally or partially occluded due to a molding
               defect, which may cause increased air volume/pressure to the
               patient and water back-up into the patient circuit.

_______________
PRODUCT        Cordis Webster Fixed-Curve Electrophysiology Catheters. 
               Recall #Z-647-5.
CODE           Sterilization lot numbers 834 through 857.
MANUFACTURER   Cordis Webster, Inc., Baldwin Park, California.
RECALLED BY    Manufacturer, by letter on or about December 30, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       5,638 catheters were distributed.
REASON         The sterility of the medical device may be compromised as
               evidenced by loss of package integrity.

_______________
PRODUCT        Fluoroplastic & Platinum Pistons, Fluoroplastic & Stainless
               Steel Pistons and Fluoroplastic Wire Pistons: 
               1.   Fluoroplastic & Platinum Piston, 0.5x4.0 mm,
                    Product No. 11-56221;
                    Fluoroplastic & Platinum Piston, 0.5x4.25 mm,
                    Product No. 11-56222;
                    Fluoroplastic & Platinum Piston, 0.5x4.50 mm
                    Product No. 11-56223;
                    Fluoroplastic & Platinum Piston, 0.5x4.75 mm
                    Product 11-56224;
               2.   Fisch Platinum Piston, 0.4x6.00 mm
                    Product No. 11-56234;
                    Fisch Platinum Piston, 0.5x6.00 mm
                    Product No. 11-56235;
                    Fisch Platinum Piston, 0.6x6.00 mm
                    Product 11-56236;
               3.   Fluoroplastic & Platinum, 0.6x4.00 mm
                    Product No. 11-56255;
                    Fluoroplastic & Platinum, 0.6x4.25 mm
                    Product No. 11-56256;
                    Fluoroplastic & Platinum, 0.6x4.50 mm
                    Product No. 11-56257;
                                    -9-                    Fluoroplastic & Platinum, 0.6x4.75 mm
                    Product No. 11-56258;
               4.   De la Cruz Platinum Piston, 0.6x3.75 mm
                    Product No. 11-56278;
                    De la Cruz Platinum Piston, 0.6x4.00 mm
                    Product No. 11-56279;
                    De la Cruz Platinum Piston, 0.6x4.25 mm
                    Product No. 11-56280;
                    De la Cruz Platinum Piston, 0.6x4.50 mm
                    Product No. 11-56281;
                    De la Cruz Platinum Piston, 0.6x4.75 mm
                    Product No. 11-56282;
                    De la Cruz Platinum Piston, 0.6x5.00 mm
                    Product No. 11-56283;
               5.   Fisch Stainless Steel Piston, 0.5x6.00 mm
                    Product No. 11-56325;
                    Fisch Stainless steel Piston, 0.6x6.00 mm
                    Product No. 11-56326;
               6.   De la Cruz Stainless Steel Piston, 0.6x3.75 mm
                    Product No. 11-56478;
                    De la Cruz Stainless Steel Piston, 0.6x4.00 mm
                    Product No. 11-56479;
                    De la Cruz Stainless Steel Piston, 0.6x4.25 mm
                    Product No. 11-56480;
                    De la Cruz Stainless Steel Piston, 0.6x4.50 mm
                    Product No. 11-56481;
                    De la Cruz Stainless Steel Piston, 0.6x4.75 mm
                    Product No. 11-56482;
               7.   Sanna Fluoroplastic - Wire Piston,
                    Product No. 91-22100;
               8.   Special Fisch Piston, Product No. 91-33155
               9.   Mass Eye & Ear Piston, 4.50 mm,
                    Product No. 91-36000;
               10.  Mass Eye & Ear Piston, 4.0 mm,
                    Product No. 91-36001;
                    Mass Eye & Ear Piston, 4.25 mm
                    Product No. 91-36002;
               11.  Fluoroplastic - Platinum Piston,
                    0.5x3.75 mm, Product No. 91-56220;
               12.  Fisch Malleus Piston, 0.4x9 mm,
                    Product No. 91-56238.  Recall #Z-652/662-5.
CODE           All lot numbers.
MANUFACTURER   Xomed Surgical Products, Jacksonville, Florida.
RECALLED BY    Manufacturer, by letter January 10, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Firm estimates 600 unimplanted units in commerce.
REASON         The wire may separate from the piston during
               surgery/manipulation of the wire to custom fit the patient's
               ear space.  

                                   -10-_______________
PRODUCT        RMI Pediatric Retrograde Coronary Sinus Perfusion Cannulae. 
               Recall #Z-663/664-5.
CODE           (a) Product Code RC-09-M, Lot numbers:  75243 and 75990; (b)
               Product Code RC-09-W, Lot #72625.
MANUFACTURER   Research Medical, Inc., Midvale, Utah.
RECALLED BY    Manufacturer, by fax March 8, 1994, followed by letter. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Florida, Texas, Virginia, Michigan, Minnesota, Maine,
               California, Oregon, Missouri, Germany, South Africa, Canada,
               Sweden, France, Israel, Greece, Japan.
QUANTITY       346 units were distributed.
REASON         Device has out of specification tip hole dimensions which
               may cause elevated cardioplegia infusion line pressures
               during the use of the cannula.

_______________
PRODUCT        Nonin Model 8800G ECG Cables, used with Nonin Model 8800
               Cardio-Respiratory Oximeters.  Recall #Z-665-5.
CODE           Cable Serial numbers:  KRE2769 through KRE3052.
MANUFACTURER   Nonin Medical, Inc., Plymouth, Minnesota.
RECALLED BY    Manufacturer, by letter February 17, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Alabama, California, Georgia, Illinois, Kentucky, Louisiana,
               Massachusetts, Minnesota, Oregon, Pennsylvania, Tennessee,
               international.
QUANTITY       63 cables.
REASON         The ECG cables are miswired which results in incorrect
               electrode placement or continuous LEAD alarm.  Misplaced
               electrodes can alter the electrocardiogram waveform.

_______________
PRODUCT        Advanced TMR Treatment System, Rx iontophoresis and
               electrotherapy device powered by rechargeable batteries.  
               Recall #Z-666-5.
CODE           A portion of the units are coded with serial numbers 1018-
               1068.
MANUFACTURER   Lake Medical Products, Inc., St. Louis, Missouri.
RECALLED BY    Manufacturer, by telephone on or about the week of January
               8, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Missouri, Kansas, New York, Ohio, Pennsylvania, South
               Carolina, Virginia, West Virginia.
QUANTITY       49 units were distributed; firm estimates none remains on
               the market.
REASON         Product was distributed without 510(k) approval.

_______________
PRODUCT        Turbo Membrane Oxygenators:  (a) Catalog #9444, Turbo
               Membrane Oxygenator for Water Delivery System, Part numbers
               98-0702-0575-6, 98-0702-0592-1; (b) Catalog #9445, Turbo
               Membrane Oxygenator with Integral Reservoir, Part numbers 
               98-0702-0591-3; (c) Catalog #9446, Turbo Membrane Oxygenator
               
                                   -11-               with Integral Cardiotomy Reservoir, Part number 
               98-0702 -577-2); (d) Catalog #4408, Holder with Water System
               for Turbos 9444, 9445, 9446; (e) Catalog #5574, Reservoir
               Support Bracket, Requires Holder 4408.  Recall #Z-667/671-5.
CODE           All lots of each product.
MANUFACTURER   Sarns, 3M Health Care, Ann Arbor, Michigan.
RECALLED BY    Manufacturer, by telephone beginning November 12, 1993, and
               by letter November 17, 1993.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       11,497 units were distributed.
REASON         Blood to air leaks at the heat exchanger seal area which
               have resulted in blood loss and subsequent oxygenator
               changeout.  

_______________
UPDATE         Recall #Z-553-5, Drake Willock System 1000 Single Patient
               Delivery System, which appeared in the April 5, 1995
               Enforcement Report should read:
               REASON:  The CPVC may swell and impede the flow meter, the
               input pressure equalizer and the output pressure equalizer,
               supply regulator, rinse block fitting, concentrate
               connectors, ultrafiltration removal regulator and dialyzer
               fittings.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        EVIS Video Endoscopes, used for endoscopic diagnosis and
               therapeutic procedures:  (a) Model GIF-130;
               (b) Model GIF-1T130; (c) Model JF-130; (d) Model TJF-130.
               Recall #Z-648/651-5.
CODE           Serial numbers included are:  
               (a) GIF-130: 
               2412211        2412385        2412445 
               2412247        2412386        2412446 
               2412248        2412388        2412447 
               2412252        2412389        2412448 
               2412254        2412390        2412449 
               2412271        2412392        2412450 
               2412272        2412393        2412451 
               2412276        2412394        2412452 
               2412281        2412395        2412453 
               2412283        2412397        2412454 
               2412285        2412400        2412455 
               2412289        2412402        2412456 
               2412292        2412403        2412457 
               2412296        2412404        2412458 
               2412302        2412407        2412459 
               2412304        2412408        2412460 
               2412314        2412409        2412461 
               2412317        2412411        2412462 

                                   -12-               2412319        2412412        2412463 
               2412320        2412413        2412464
               2412323        2412414        2412465 
               2412324        2412415        2412466 
               2412325        2412416        2412467 
               2412326        2412417        2412468 
               2412327        2412418        2412469 
               2412335        2412419        2412470 
               2412337        2412422        2412471 
               2412338        2412423        2412472 
               2412339        2412424        2412473
               2412340        2412425        2412474 
               2412349        2412426        2412475 
               2412354        2412427        2412476 
               2412357        2412428        2412478 
               2412362        2412429        2412479 
               2412364        2412430        2412480 
               2412365        2412431        2412481 
               2412367        2412432        2412482 
               2412368        2412434        2412483 
               2412371        2412435        2412484 
               2412373        2412436        2412485 
               2412374        2412437        2412486 
               2412375        2412438        2412487 
               2412376        2412439        2412488 
               2412377        2412440        2412489 
               2412378        2412441        2412490 
               2412381        2412442        2412491 
               2412382        2412443        2412492 
               2412383        2412444        2412493 
               2412384        2412445        2412494
               2412495        2412559        2412649 
               2412496        2412560        2412654 
               2412497        2412561        2412655 
               2412498        2412562        2412659 
               2412499        2412563        2412676 
               2412500        2412566        2412681 
               2412501        2412568        2412682 
               2412502        2412569        2412685 
               2412503        2412573        2412689 
               2412504        2412574        2412690 
               2412505        2412575        2412696 
               2412506        2412576        2412698 
               2412507        2412577        2412699 
               2412508        2412578        2412700 
               2412509        2412581        2412701 
               2412511        2412582        2412702 
               2412512        2412583        2412703 
               2412514        2412585        2412704
               2412515        2412586        2412706 
               2412516        2412587        2412710 
               2412517        2412588        2412711 
               2412518        2412590        2412713 
                                   -13-               2412519        2412592        2412715 
               2412520        2412600        2412716 
               2412521        2412602        2412717 
               2412522        2412604        2412718 
               2412524        2412605        2412720 
               2412526        2412606        2412722 
               2412527        2412607        2412723 
               2412529        2412608        2412724 
               2412531        2412609        2412725 
               2412533        2412611        2412726 
               2412534        2412612        2412728 
               2412535        2412615        2412729 
               2412536        2412616        2412732 
               2412538        2412619        2412738 
               2412539        2412621        2412739 
               2412541        2412622        2412742 
               2412543        2412623        2412753 
               2412544        2412624        2412757 
               2412546        2412625        2412758 
               2412547        2412626        2412760 
               2412549        2412627        2412765
               2412551        2412632        2412767 
               2412552        2412634        2412768 
               2412554        2412639        2412773 
               2412555        2412642        2412776 
               2412556        2412644        2412777 
               2412557        2412646        2412781 
               2412558        2412482        2412786 
               2412788        2412790        2412792 
               2412797        2412798        2412800 
               2412803        2412805        2412807 
               2412809        2412815        2412816 
               2412817        2412820        2412821 
               2412822        2412826        2412831 
               2412833        2412834        2412835 
               2412836        2412838        2412840 
               2412845        2412846        2412848 
               (b) GIF-1T30
               2400164        2400173        2400175
               2400177        2400179        2400182
               2400185        2400187        2400189     2400191  
               (c)  JF-130 
               2410484        2410494        2410511     2410512
               2410518        2410524        2410525     2410529 
               2410532        2410534        2410538     2410541  
               2410542        2410543        2410544     2410546 
               2410547        2410548        2410549     2410550
               2410551        2410553        2410554     2410555
               2410558        2410560        2410561     2410562
               2410563        2410566        2410567     2410568
               2410571        2410572        2410573     2410574
               2410575        2410576        2410577     2410578
               2410579        2410580        2410581     2410582 
                                   -14-               2410590        2410593        2410597     2410599
               2410600        2410604
               (d)  TJF-130:                              
               2400111        2400116        2400117     2400118
               2400123        2400126        2400133     2400135
               2400145        2400147        2400148     2400149
               2400151        2400153        2400154     2400155 
               2400156        2400158        2400159     2400161 
               2400162        2400164        2400166     2400167 
               2400168        2400169        2400172     2400173
               2400174        2400176        2400177     2400178
               2400179        2400180        2400181     2400182
               2400183        2400185. 
MANUFACTURER   Olympus Optical Co., Ltd., Tokyo, Japan 
RECALLED BY    Olympus America Inc., Lake Success, New York, by letters of
               February 8, 1995 and March 27, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide, Canada, South American, Central America.
QUANTITY       Approximately 422 units are affected by this recall.
REASON         The video endoscopes are more susceptible to surges of
               electrical current due to a manufacturing defect in the
               charged coupled device (CCD) chip which will cause the video
               endoscopes to lose their images.

_______________
PRODUCT        Candela PLDL and PLTL Dermatology Laser Systems, used to
               treat vascular lesions, brown lesions, and tattoos.  
               Recall #Z-672/679-5. 
CODE           1.  PLDL 9914-00-0200: serial no. 9914-0200-0139, 0153,
               0175, 0181, 0182, 0192, 0193, 0194, 0195, 0198, 0199, 0200,
               0202, 0204, 0206, 0207, 0209, 0210, 0216, 0217, 0221, 0223
               2.   PLTL 9914-00-0100: serial no. 9914-0100-4092
               3.   PLTL 9914-00-0150: serial no. 9914-0150-4101, 4140
               4.   PLTL 9914-00-0151: serial no. 9914-0151-4126, 4138
               5.   PLTL 9914-00-0152: serial no. 9914-0152-4098, 4100,
               4104, 4115, 4116, 4118, 4119, 4121, 4125, 4127, 4132, 4135,
               4136, 4144
               6.  PLTL 9914-00-1000: serial no. 9914-1000-4152
               7.  PLTL 9914-00 1010: serial no. 9914-1010-4082, 4097,
               4106, 4129, 4131, 4134, 4137, 4141, 4142, 4147, 4148, 4149
               8.  PLTL 9914-00-1110: serial no. 9914-1110-4113, 4154,
               4156, 4158 - 4174, 4176 - 4180, 4183, 4184, 4186 - 4191,
               4196, 4197, 4201, 4203, 4205, 4208, 4211 - 4215, 4218 -
               4220, 4222. 
MANUFACTURER   Candela Laser Corporation, Wayland, Massachusetts.
RECALLED BY    Manufacturer, by letter May 13, 1994.  Firm-initiated field
               correction complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       All 94 products on the market have been upgraded and all 8
               products in house have also been upgraded.
REASON         The design allowed for possible internal fires.  


                                   -15-RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
===========
_______________
PRODUCT        Kent MAX-E-CAL SOW 27 NT Medicated Type B Feed, in 50 pound
               bags.  Recall #V-028-5.
CODE           Lot RA126 and R126.
MANUFACTURER   Kent Feeds, Inc., Rockford, Illinois.
RECALLED BY    Manufacturer, by telephone October 19 and 26, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, Wisconsin.
QUANTITY       396 bags were distributed.
REASON         Feed was subpotent for neomycin base.


MEDICAL DEVICE SAFETY ALERT: 
=============================================
_______________
PRODUCT        Servo 300 Ventilator.  Safety Alert #N-026-5.
CODE           Product #6079391 E338E.
MANUFACTURER   Siemens-Elema AB, Solna, Sweden.
ALERTED BY     Siemens Medical Systems (SMS), Inc., Danvers, Massachusetts,
               by letter March 7, 1995.
DISTRIBUTION   Nationwide.
QUANTITY       161 units.
REASON         When the product is exposed to high levels of electrostatic
               or electromagnetic interference the device may exhibit
               functional irregularities.

                                   -16-

END OF ENFORCEMENT REPORT FOR APRIL 19, 1995.  BLANK PAGES MAY