Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, Department of Health and Human Services. It contains information on actions taken in connection with agency Regulatory activities.

September 4, 2002
02-35

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS I

_______________________
PRODUCT
SUPREME QUALITY brand PICKLED Sour bamboo shoot (SLICE) packed in glass jars, NET WT. 29.89 OZ. ING: Bamboo, Water, Salt. PRODUCT OF THAILAND. Barcode #84909 00343. Recall # F-746-2.
CODE
All products on the market at the time the recall was initiated.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Hong Thai Foods Corp., Brooklyn, NY, by press release on February 14, 2002.
Manufacturer: Savoy Enterprises Ltd., Part. Yannawa, Bangkok.
State initiated recall is complete.
REASON
The New York State Department of Agriculture and Markets found the product to contain undeclared sulfites (390 ppm).
VOLUME OF PRODUCT IN COMMERCE
23 cases (12 jars per case).
DISTRIBUTION
NY and NJ.

_______________________
PRODUCT
Caravelle brand, Pickled Sour Bamboo Shoot Slices, 2 pound and 4 pound jars. Recall # F-747-2.
CODE
All products on the market at the time the recall was initiated.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Anhing Corporation, Los Angeles, CA, by letter on July 2, 2002. The New York State Department of Agriculture and Markets issued a press release on June 27, 2002.
Manufacturer: Unknown. State initiated recall is ongoing.
REASON
The New York State Department of Agriculture and Markets found the product to contain 781 ppm undeclared sulfites in 35g serving size.
VOLUME OF PRODUCT IN COMMERCE
Undetermined.
DISTRIBUTION
Nationwide.

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS II

_______________________
PRODUCT
Pure Fun brand Peach Mango Juice Blend Concentrate,1+7, in half gallon bottles. Recall # F-748-2.
CODE
Lot 110702.
RECALLING FIRM/MANUFACTURER
Penguin Juice Company, Westland, MI, by telephone on July 29, 2002 and by letter on August 1, 2002. FDA initiated recall is complete.
REASON
The product contained undeclared FD&C Yellow No. 5.
VOLUME OF PRODUCT IN COMMERCE
50 cases (6 bottles per case).
DISTRIBUTION
GA.

_______________________
PRODUCT
Laima brand Greipfrutu Skelites (Grapefruit Candy). Recall # F-749-2.
CODE
Lot No. L01012.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Romic Inc., Alpharetta, GA, by telephone on June 13, 2002.
Manufacturer: JSC Laima Riga, Latvia. State initiated recall is ongoing.
REASON
The product contained the unapproved food color Ponceau 4R.
VOLUME OF PRODUCT IN COMMERCE
640 cartons.
DISTRIBUTION
NY.

_______________________
PRODUCT
Miguel and Valentino brand Fig Cake, Hand Made, packaged in cellophane wrap and further packaged in wraparound wicker basket, net wt. approx 2 lbs. Recall # F-750-2.
CODE
Lots 14922, 23206, 25093, 27205.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Miguel and Valentino, Bethesda, MD, by press release on June 6, 2002.
Manufacturer: Victoriano Navas Martin Zorilla Archez (Malaga) Spain. State initiated recall is complete.
REASON
Product contained undeclared almonds which were visible on the top of the cake.
VOLUME OF PRODUCT IN COMMERCE
672 cakes.
DISTRIBUTION
CO, FL, GA, MD, D.C., MI, CA and NJ.

RECALLS AND FIELD CORRECTIONS: DRUGS -- CLASS II

_______________________
PRODUCT
Zerit (stavudine) Capsules, 40 mg, bottles of 60, Rx only. Recall # D-398-2.
CODE
Lot 1A36209, Exp. FEB 2003.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Alliance Wholesale Distributor, Richton Park, IL, by letter dated June 28, 2002.
Manufacturer: Bristol-Myers Squibb Oncology/Immunology, Princeton, NJ. Firm initiated recall is ongoing.
REASON
Tablets changed to capsules.
VOLUME OF PRODUCT IN COMMERCE
26 bottles.
DISTRIBUTION
Nationwide.

_______________________
PRODUCT
Quinaglute (quinidine gluconate, USP), Dura-Tabs (extended-release tablets, USP) 324 mg, 100 count bottles, Rx only. Recall # D-399-2.
CODE
Lot #012634, Exp. OCT 23 03
Lot #012831, Exp. NOV 27 03
Lot #012263, Exp. SEP 04 03.
RECALLING FIRM/MANUFACTURER
Recalling Firm: AmeriSource Health Services Corp., Columbus, OH, by letter and fax on August 20, 2002.
Manufacturer: Berlex Laboratories, Wayne, NJ. Firm initiated recall is ongoing.
REASON
Metal particles found in tablet (by manufacturer).
VOLUME OF PRODUCT IN COMMERCE
687 bottles total.
DISTRIBUTION
Nationwide.

_______________________
PRODUCT
Nasonex Nasal Spray (mometasone furoate monohydrate) 50 mcg, Net Contents: 17 g, 120 Metered Sprays, FOR INTRANASAL USE ONLY. Rx only. Recall # D-400-2.
CODE
1-KTL-147 Exp. date 3.03.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Schering Corpration, Kenilworth, NJ, by fax on an unknown date and by letters on August 12, 2002.
Manufacturer: Schering-Plough Products, Inc., Manati, Puerto Rico. Firm initiated recall is ongoing.
REASON
Defective container; missing part in actuator may cause a narrow stream of product to be dispensed rather than a fine spray
VOLUME OF PRODUCT IN COMMERCE
295,620 units.
DISTRIBUTION
Nationwide.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS I

_______________________
PRODUCT
Human Tissue for Transplantation of the following types. Recall # B-1789-2.
a) Skin, Meshed
b) Skin (sq.ft.)
c) Tibia
d) Achilles Tendon with Bone
e) Patellar Ligament; with Quadriceps
f) Patellar Ligament
g) Ilium, Hemi
h) Fibula, Segment
i) Tri-Cortical Block
J) Cortical Wedge Select
k) Fibula, Ring
l) Cancellous, Crushed
m) Cancellous Block
n) Cancellous, Crushed
CODE
a) Tissue ID # CTSP000348-009, CTSP000348-010,
CTSP000348-011, CTSP000348-012,
CTSP000348-013, CTSP000348-014.
b) Tissue ID # CTSP000348-015, CTSP000348-016.
c) Tissue ID # CTSP000348-017.
d) Tissue ID # CTSP000348-018, SP000348-019.
e) Tissue ID # CTSP000348-020.
f) Tissue ID # CTSP000348-021.
g) Tissue ID # CTSP000348-022.
h) Tissue ID # CTSP000348-023, CTSP000348-024. i) Tissue ID # CTSP000348-032.
j) Tissue ID # CTSP000348-033, CTSP000348-034,
CTSP000348-035, CTSP000348-036, CTSP000348-037, CTSP000348-038, CTSP000348-039, CTSP000348-040, CTSP000348-041.
k) Tissue ID # CTSP000348-044, CTSP000348-046, CTSP000348-050.
l) Tissue ID # CTSP000348-053, CTSP000348-054,
CTSP000348-055, CTSP000348-056,
CTSP000348-057, CTSP000348-058,
CTSP000348-059, CTSP000348-060, CTSP000348-061, CTSP000348-062.
m) Tissue ID # CTSP000348-063, CTSP000348-064, CTSP000348-065, CTSP000348-066, CTSP000348-067, CTSP000348-068, CTSP000348-069, CTSP000348-070, CTSP000348-071, CTSP000348-072, CTSP000348-073, CTSP000348-074, CTSP000348-075, CTSP000348-076,
CTSP000348-077, CTSP000348-078,
CTSP000348-079.
n) Tissue ID # CTSP000348-080, CTSP000348-081, CTSP000348-082.
RECALLING FIRM/MANUFACTURER
Community Blood Center, dba: Community Tissue Services (CTS), Dayton, Ohio, by letters dated July 15, 2002 and July 25, 2002.
Firm initiated recall is ongoing.
REASON
Tissue for transplantation, processed from a donor who tested positive for hepatitis C virus (HCV) subsequent to distribution.
VOLUME OF PRODUCT IN COMMERCE
59 units.
DISTRIBUTION
Nationwide.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II

_______________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Recall # B-1788-2.
CODE
Units 04KS64220, 04G25447.
RECALLING FIRM/MANUFACTURER
American Red Cross, New England Region, Dedham, MA, by letter dated May 7, 2002. Firm initiated recall is complete.
REASON
Blood products, collected from a donor who had lived in an area designated as endemic for malaria, were distributed.
VOLUME OF PRODUCT IN COMMERCE
2 units.
DISTRIBUTION
MA.

_______________________
PRODUCT
Platelets, Pheresis, Leukocytes Reduced, Recall # B-1818-2
CODE
Unit 04GS06136.
RECALLING FIRM/MANUFACTURER
American Red Cross Blood Services, New England Region, Dedham, MA, by telephone on May 13, 2002, and letter dated May 28, 2002.
Firm initiated recall is complete.
REASON
Blood product, inappropriately prepared in that the product was distributed in one platelet bag instead of two, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
NY.

_______________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Irradiated, Recall # B-1820-2.
CODE
Unit S12185.
RECALLING FIRM/MANUFACTURER
Delta Blood Bank, Stockton, CA, by letter dated November 2, 2000.
Firm initiated recall is complete.
REASON
Blood product, which tested non-reactive for antibody to the hepatitis B core antigen (anti-HBcore), but was collected from a donor who previously tested reactive on two prior occasions, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
CA.

_______________________
PRODUCT
Source Plasma, Recall # B-1833-2.
CODE
Units 0015331, 0015469, 0015654.
RECALLING FIRM/MANUFACTURER
DCI Biologicals Hot Springs, Inc., Plasmapheresis Facility, Hot Springs AK, by fax on February 11, 2002. Firm initiated recall is complete.
REASON
Blood products, collected from a donor who tested reactive for hepatitis B surface antigen (HbsAg), were distributed.
VOLUME OF PRODUCT IN COMMERCE
3 units.
DISTRIBUTION
Austria.
______________________
PRODUCT
Red Blood Cells, Leukoreduced, Recall # B-1834-2.
CODE
Units 50C19470, 50C19479, 50C19554, 50LG55918, 50C19378, 50C19350, 50LG55838, 50C19960, 50C19973, 50C19976, 50C20073, 50C23477, 50LG58467, 50LG58471, 50LG58483, 50LG57697, 50C23668, 50C23712, 50C23713, 50LH24269, 50LJ36051, 50LJ36210, 50LJ36248, 50C24394.
RECALLING FIRM/MANUFACTURER
American Red Cross Blood Services, Western Lake Erie Region, Toledo, OH, by letter dated December 4, 2001 or February 6, 2002.
Firm initiated recall is complete.
REASON
Blood products, which were incorrectly tested for Cytomegalovirus (CMV), were distributed.
VOLUME OF PRODUCT IN COMMERCE
24 units.
DISTRIBUTION
OH.

_______________________
PRODUCT
Platelets Pheresis, Recall # B-1857-2.
CODE
Unit number R52990.
RECALLING FIRM/MANUFACTURER
HCSC Blood Center, Miller Memorial Blood Center, Bethlehem, PA, by letter on August 8, 2000. Firm initiated recall is complete.
REASON
Platelets with a decreased platelet count were distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
PA.

_______________________
PRODUCT
Platelets Pheresis, Recall # B-1858-2.
CODE
Unit number R52937.
RECALLING FIRM/MANUFACTURER
HCSC Blood Center, Miller Memorial Blood Center, Bethlehem, PA, by letter on July 20, 2000. Firm initiated recall is complete.
REASON
Blood product, collected from a donor that did not answer one of the medical history questions concerning behavior known to increase risk of infection with the human immunodeficiency virus (HIV), was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
PA.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS III

_______________________
PRODUCT
Source Plasma, Recall # B-1707-2.
CODE
Units 46 DCDYDG 0, 46 DCCWXG 0.
RECALLING FIRM/MANUFACTURER
Aventis Bio-Services, Inc., Decatur, IL, by fax on February 22, 2002.
Firm initiated recall is complete.
REASON
Blood products, collected from a donor with an elevated temperature and whose health history was incomplete, were distributed.
VOLUME OF PRODUCT IN COMMERCE
2 units.
DISTRIBUTION
IL.

_______________________
PRODUCT
Platelets, Pheresis, Recall # B-1816-2.
CODE
Unit R79836.
RECALLING FIRM/MANUFACTURER
HCSC Blood Center, t/a Miller Memorial Blood Center, Bethlehem, PA, by telephone on August 6, 2001 and by letter dated September 14, 2001. Firm initiated recall is complete.
REASON
Blood product, collected from a donor whose arm inspection was not documented, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
PA.

_______________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Recall # B-1819-2.
CODE
Unit 12GY49326.
RECALLING FIRM/MANUFACTURER
American Red Cross Blood Services, Carolinas Region, Charlotte, NC, by telephone on May 29, 2002 and by letter dated June 11, 2002. Firm initiated recall is complete.
REASON
Blood product, which had a positive antibody screening test, but labeled as negative for unexpected antibodies, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
NC.

_______________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Recall # B-1821-2.
CODE
Unit G46085.
RECALLING FIRM/MANUFACTURER
Northern Illinois Blood Bank, Rockford, IL, by telephone on June 6, 2002. Firm initiated recall is complete.
REASON
Blood product, which was incorrectly tested for anti-HTLV I/II, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
IL.

_______________________
PRODUCT
Fresh Frozen Plasma, Recall # B-1855-2.
CODE
Unit number 6745504.
RECALLING FIRM/MANUFACTURER
Coral Blood Services, Inc., Children's Memorial Hospital, Chicago, IL, by telephone on February 22, 2002 and by letter dated March 27, 2002. Firm initiated recall is complete.
REASON
Blood product, that was not frozen within 8 hours of collection, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit.
DISTRIBUTION
IL.

_______________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Recall # B-1856-2.
CODE
Unit numbers 6745708, 6746142, 6746143, 6746145, 6746146, and 6746147.
RECALLING FIRM/MANUFACTURER
Coral Blood Services, Inc., Children's Memorial Hospital, Chicago, IL, by telephone on February 21, 2002 and by letter dated March 27, 2002. Firm initiated recall is complete.
REASON
Blood products, that did not have the additive solution included within the time frames specified by the manufacturer's instructions, were distributed.
VOLUME OF PRODUCT IN COMMERCE
6 units.
DISTRIBUTION
IL.

_______________________
PRODUCT
Red Blood Cells, Recall # B-1859-2.
CODE
Unit numbers T30834, T30836, and T30841.
RECALLING FIRM/MANUFACTURER
HCSC Blood Center, Miller Memorial Blood Center, Bethlehem, PA, by telephone on April 19, 2000. Firm initiated recall is complete.
REASON
Unlicensed blood products were distributed interstate.
VOLUME OF PRODUCT IN COMMERCE
3 units.
DISTRIBUTION
NJ.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

****CORRECTION****
Recall # Z-1152-2/Z-1154-2 on Enforcement Report 02-30 had the incorrect manufacturer listed. The correct manufacturer is: Angeion Corp., Minneapolis, MN.

_______________________
PRODUCT
Sterile Laceration Tray, product code 50-10890, packaged in sterile tray, 30 trays\case. Recall # Z-1269-2.
CODE
Lot number 324580.
RECALLING FIRM/MANUFACTURER
Recalling Firm: Deroyal Industries, Inc., Powell, TN, by letter on March 19, 2002.
Manufacturer: Deroyal Wound Care, Rose Hill, VA. Firm initiated recall is complete.
REASON
Defective seals on surgical trays compromise device sterility.
VOLUME OF PRODUCT IN COMMERCE
90 trays.
DISTRIBUTION
GA.

_______________________
PRODUCT
Buffered Iontophoretic Delivery Electrode System, non-sterile, single-use only, packaged in a metalized mylar bag, 12 per box.
IO-Drive Iontophoresis Electrode System. Recall # Z-1271-2.

The product consists of 1 active drug delivery electrode, 1 return electrode, 1 or 2 alcohol swabs, and an instruction manual. The device is sold as a kit that is non-sterile and intended for single-use only.
The kits contain one of three sizes of the active electrode (small, medium, or large).
The medical device is sold under the 'Selective Med Components' label and several private labels as follows:
1) Buffered Iontophoretic Delivery Electrode System; North Coast Medical, Inc.
2) Buffered Iontophoretic Delivery Electrode System, EMSI, Tampa, Florida.
3) Buffered Iontophoretic Delivery Electrode System; MedPend, Inc., Lake Mary, FL.
4) Buffered Iontophoretic Delivery Electrode System; Rudolph Medical.
5) Buffered Iontophoretic Delivery Electrode System; Physio Tech.
6) Buffered Iontophoretic Delivery Electrode System;
Spectramed.
7) IoDOSE Buffered Iontophoretic Delivery Electrode System;
MagMed, Inc.
8) SUPER-TRODES Buffered Iontophoretic Delivery Electrode
System; SME, Inc.
9) Buffered Iontophoretic Delivery Electrode System; Selective
Med Components.
10) IONTO-Clone Buffered Iontophoretic Delivery Electrode System; New England Surgical, Inc.
CODE
Lot Numbers: 5129-1, 5130-1, 5131-1, 5291-1, 5292-1, 5293-1, 5368-1, 5369-1, 5370-1, 5573-1, 5574-1, 5575-1, 5684-1, 5685-1, 5686-1, 5946-1, 5947-1, 5948-1, 6483-1, 6484-1, 6487-1, 6711-1, 6712-1, 6713-1, 7030-1, 7031-1, 7032-1, 7427-1, 7428-1, 7429-1, 7513-1, 7514-1, and 7515-1.
RECALLING FIRM/MANUFACTURER
Selective Med Components, Inc., Mt. Vernon, OH, by recall notice on July 18, 2002. FDA initiated recall is ongoing.
REASON
The device may cause skin burns.
VOLUME OF PRODUCT IN COMMERCE
70,000 kits.
DISTRIBUTION
FL, MS, OH, MA, CA, VA, and NC.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III

_______________________
PRODUCT
Storz brand Concentrix Phacoemulsification Pack, Labeled as Sterile. Product or reference number CX4310. Recall # Z-1270-2.
CODE
Lot Number: S7825.
RECALLING FIRM/MANUFACTURER
Bausch & Lomb Surgical, Inc., Clearwater, FL, by letter on May 10, 2002. Firm initiated recall is ongoing.
REASON
Irrigation line tubing may not properly adhere to the casette and leak or detach.
VOLUME OF PRODUCT IN COMMERCE
600 (100 boxes/6 units each).
DISTRIBUTION
Nationwide and Australia, Canada and The Netherlands.

END OF ENFORCEMENT REPORT FOR SEPTEMBER 4, 2002

####