FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, Department of Health and Human Services. It contains information on actions taken in connection with agency Regulatory activities.


July 04, 2001 01-26

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS III

_______________________________ RECALL NUMBER, PRODUCT AND CODE: F-435-1 Product labeled as: "Nature's Wealth brand Chelated Potassium 99 mg Dietary Supplement", 100-tablets bottles. Firm identified on the label: "Natural Welath Nutrition Corp Bohemia, Ny". REASON: Bottles labeled as "Chelated Potassium 99 mg" actually contained a different product, "Ultra Vita-Min Ultra Potency Vitamin & Mineral Formula". MANUFACTURER: NBTY, Inc., Bohemia, NY RECALLED BY: The firm sent a recall letter, dated 12/01/00. FIRM INITIATED RECALL: Complete. DISTRIBUTION: Nationwide QUANTITY: 1086 - 100 tablet bottles

RECALLS AND FIELD CORRECTIONS: DRUGS -- CLASS II

________________________________ RECALL NUMBER, PRODUCT AND CODE: D-240-1, ADAGEN (pegademase bovine injection) 250 units per mL. Intramuscular use only 1.5 mL single-dose vials (units of 4) Rx only. NDC 57665-001-01 NDA 19-818 REASON: Cross contamination with FOMEPIZOLE (4-METHYL-PYRAZOLE) MANUFACTURER/RECALLING FIRM: Enzon, Inc, Piscataway, NJ RECALLED BY: Manufacturer, on March 23, 2001, by telephone. FIRM INITIATED RECALL: Ongoing DISTRIBUTION: CA, FL, IL, LA, MD, MA, NY, NC, OH, OK, TN, TX, UT and including Japan, Australia, Canada, Denmark, Croatia, France, Germany, Italy, Switzerland, and Spain QUANTITY 792 vials _______________________________ RECALL NUMBER, PRODUCT AND CODE: D-241-1, Serostim, 6 mg. vials (somatropin rDNA origin for injection), 7 vials per carton with 7 vials of sterile diluent. Lot Number MNH605A, Exp. 4/02 REASON: COUNTERFEIT MANUFACTURER: Unknown RECALLED BY: Distributor or recalling firm, issued recall notice via letter on 5/15/01. FIRM INITIATED RECALL: Ongoing DISTRIBUTION: GA, SC and SC QUANTITY: 4 cartons, 7 x 6 mg. vials per carton

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II

_______________________________ RECALL NUMBER, PRODUCT AND CODE: B-887-1, Red Blood Cells, unit 4056710 B-888-1, Platelets, unit 4056710 REASON: Blood products, collected from a donor with Crohn’s disease, were distributed. MANUFACTURER/RECALLING FIRM: Community Blood Center, Dayton, OH RECALLED BY: The recalling firm, by letters dated September 28, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: OH QUANTITY: 2 units _______________________________ RECALL NUMBER,PRODUCT AND CODE: B-1130-1, Red Blood Cells Leukoreduced, unit 12GR03564 REASON: Blood product, which was collected from a donor with active post- poliomyelitis muscular atrophy, was distributed. MANUFACTURER/RECALLING FIRM: American Red Cross, Carolinas Region, Charlotte, NC RECALLED BY: The recalling firm, by telephone on February 18, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: NC QUANTITY: 1 unit ________________________________ RECALL NUMBER, PRODUCT AND CODE: B-1449-1, Red Blood Cells, unit 1727888. REASON: Blood products, collected from a donor with a history of Hodgkin’s Disease, were distributed. MANUFACTURER/RECALLING FIRM: Central Indiana Regional Blood Center, Indianapolis, Indiana RECALLED BY: The recalling firm, by letter dated September 8, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: IN and Switzerland QUANTITY 1 unit ________________________________ RECALL NUMBER, PRODUCT AND CODE: B-1461-1, Red Blood Cells, unit number 04R64529 REASON: Blood product, collected from an unsuitable donor based on living in an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, New England Region, Dedham, MA RECALLED BY: The recalling firm, by letter dated January 22, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MA QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1462-1, Red Blood Cells (2 units), unit number 04GJ02418 (split unit) REASON: Blood products, collected from an unsuitable donor based on travel to an area considered endemic for malaria, were distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, New England Region, Dedham, MA RECALLED BY: The recalling firm, by letter dated January 23, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MA QUANTITY 2 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1463-1, Red Blood Cells, unit number 04FC79460 REASON: Blood product, collected from an unsuitable donor based on living in an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, New England Region, Dedham, MA RECALLED BY: The recalling firm notified the consignee, by letter dated January 24, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MA QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1464-1, Red Blood Cells, unit number 53H77398 REASON: Blood product, collected from an unsuitable donor based on travel to an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Greater Chesapeake and Potomac Region, Baltimore, MD RECALLED BY: The recalling firm, by letters dated December 8, 2000, and December 14, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: MD QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1465-1, Red Blood Cells, Leukocytes Reduced, unit number 29GX12243 REASON: Blood product, collected from an unsuitable donor based on living in an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Mid-Atlantic Region, Norfolk, VA RECALLED BY: The recalling firm, by letter dated January 29, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: NC QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1466-1, Red Blood Cells, unit number 29FW40076 REASON: Blood product, collected from an unsuitable donor based on living in an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Mid-Atlantic Region, Norfolk, VA RECALLED BY: The recalling firm, by letter dated November 16, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: VA QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1467-1, Red Blood Cells, Leukocytes Reduced, unit number 49GN04187 REASON: Blood product, collected from an unsuitable donor based on travel to an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Southwest Region, Ponca City, OK RECALLED BY: The recalling firm, by letter dated January 8, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: OK QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1497-1, Red Blood Cells, units 12FQ48437, 12FP34499, 12FP34084, and 12FR52236 B-1498-1, Fresh Frozen Plasma, unit 12FP34084 B-1499-1, Cryoprecipitated AHF, unit 12FQ48437 B-1500-1, Recovered Plasma, units 12FQ48437, 12FP34499, 12FQ44113, and 12FR52236 REASON: Blood products, which were collected from a donor with a history of Behcet’s Syndrome, were distributed. MANUFACTURER/RECALLING FIRM: American Red Cross, Charlotte, NC RECALLED BY: The recalling firm, by telephone on June 4, 1999 or by letter dated June 8, 1999. FIRM INITIATED RECALL: Complete DISTRIBUTION: NC and Switzerland QUANTITY 10 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1506-1, Red Blood Cells, units 1233721, 1242026, 1246035, 1256941 B-1507-1, Platelets, units 1233721, 1246035 B-1508-1. Plasma, unit 1242026 B-1509-1, Fresh Frozen Plasma, unit 1246035 B-1510-1, Red Blood Cells, Leukocytes Reduced, unit 1274972 B-1511-1, Recovered Plasma, unit 1256941 REASON: Blood products, collected from an unsuitable donor due to previously testing positive for HIV, were distributed. MANUFACTURER/RECALLING FIRM: Community Blood Center of Greater Kansas City, Kansas City, MO RECALLED BY: The recalling firm, by letters dated December 6, 1999. FIRM INITIATED RECALL: Complete DISTRIBUTION: MO, KS, CA QUANTITY 10 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1523-1, Red Blood Cells (6 units), unit numbers 2201304, 8071111, 5032285, 5025214, 5016026, and 5005762 B-1524-1, Platelets (2 units), unit numbers 2201304 and 5032285 B-1525-1. Cryoprecipitated AHF (2 units), unit numbers 8071111 and 5025214 B-1526-1, Fresh Frozen Plasma (3 units), unit numbers 2201304, 5016026, and 5005762 B-1527-1, Plasma, Cryoprecipitate Removed (1unit), unit number 8071111 B-1528-1, Recovered Plasma (2 units), unit numbers 5032285 and 5025214 REASON: Blood products, that tested negative for hepatitis, but were collected from an unsuitable donor based on previous reactive hepatitis testing, were distributed. MANUFACTURER/RECALLING FIRM: LifeShare, Inc., Elyria, OH RECALLED BY: The recalling firm notified consignees in Ohio, Florida, and Tennessee. FIRM INITIATED RECALL: Complete DISTRIBUTION: OH, FL, TN QUANTITY 16 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1513-1, Red Blood Cells, Leukocytes Reduced, units 0344150, 0345460, 0347722, 0347726, 0347733, 0347738 REASON: Blood products, which were labeled as leukoreduced but did not undergo leukoreduction, were distributed. MANUFACTURER/RECALLING FIRM: Community Blood Center of Greater Kansas City, Kansas City, MO RECALLED BY: The recalling firm, by telephone on October 13, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: KA QUANTITY 6 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1514-1, Platelets Pheresis, Leukocytes Reduced (2 units), unit number 0079513 (split unit - 2 aliquots distributed) B-1515-1, Platelets Pheresis, Leukocytes Reduced Irradiated (3 units), unit numbers 0079516 (split unit - 1 aliquot distributed) and 0079518 (split unit - 2 aliquots distributed) REASON: Blood products, that were labeled leukoreduced but were not tested to determine the white blood cell count as required in the firm's standard operating procedures, were distributed. MANUFACTURER/RECALLING FIRM: Michigan Community Blood Center, Grand Valley Blood Program, Grand Rapids, MI RECALLED BY: The recalling firm, by letter on January 2, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI QUANTITY 5 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1529-1, Platelet, Pheresis, Leukoreduced , unit 3872596 REASON: Blood product, with a low platelet count, was distributed. MANUFACTURER/RECALLING FIRM: Gulf Coast Regional Blood Center, Houston, Texas RECALLED BY: The recalling firm, by fax dated June 30, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: TX QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1537-1, Red Blood Cells, unit number 3809631 B-1538-1, Cryoprecipitated AHF, unit number 3809631 B-1539-1, Plasma, Cryoprecipitate Removed, unit number 3809631 REASON: Blood products, collected from a donor who should have been deferred due to use of the medication Arthrotec, were distributed. MANUFACTURER/RECALLING FIRM: Gulf Coast Regional Blood Center, Houston, Texas RECALLED BY: The recalling firm, by letter on July 7, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: TX QUANTITY 3 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1540-1, Red Blood Cells, unit numbers 42W06650 and 42M97944 B-1541-1, Red Blood Cells, Leukocytes Reduced, unit number 42FH96660 B-1542-1, Platelets, unit number 42FH96660 REASON: Blood products, that tested negative for antibody to hepatitis B core antigen (anti-HBc), but were collected from a donor that previously tested repeatedly reactive for anti-HBc on more than one occasion, were distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Northern Ohio Region, Cleveland, OH RECALLED BY: The recalling firm, by letter dated February 26, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: OH QUANTITY 4 units. _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1543-1, Red Blood Cells, Leukocytes Reduced, unit number 18GG34166 REASON: Blood product, that tested negative for antibody to hepatitis B core antigen (anti-HBc), but was collected from a donor that previously tested repeatedly reactive for anti-HBc on more than one occasion, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Great Lakes Region, Lansing, MI RECALLED BY: The recalling firm, by letter dated July 25, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1544-1, Red Blood Cells, Leukocytes Reduced, unit numbers 18FX26150 and 18FX26162 B-1545-1, Platelets, unit numbers 18FX26150, 18FX26158, 18FX26161, and 18FX26162 REASON: Blood products, collected from donors whose donor suitability determinations were inadequately performed, were distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Great Lakes Region, Lansing, MI. RECALLED BY: The recalling firm, by telephone on January 18, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI and MN QUANTITY 6 _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1585-1, Red Blood Cells, Leukocytes Reduced, unit number 13GQ40747 B-1586-1, Platelets, unit number 13GQ40747 REASON: Blood products, collected from an unsuitable donor based on travel to an area considered endemic for malaria, were distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Southeastern Michigan Region, Detroit, MI RECALLED BY: The recalling firm, by letter dated February 2, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI QUANTITY 2 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1587-1, Platelets, unit number C24942 REASON: Blood product, collected from an unsuitable donor based on travel to an area considered endemic for malaria, was distributed. MANUFACTURER/RECALLING FIRM: Aurora Area Blood Bank, Heartland Blood Centers, Aurora, IL RECALLED BY: The recalling firm, by letter dated March 19, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: IL QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1588-1, Red Blood Cells (4 units), unit numbers 3879151 and 3758915 (split units) B-1589-1 Platelets Pheresis (1 unit), unit number 3850090 REASON: Blood products, collected from an unsuitable donor based on travel to an area considered endemic for malaria, were distributed. MANUFACTURER/RECALLING FIRM: Gulf Coast Regional Blood Center, Houston, TX RECALLED BY: The recalling firm, by letters on October 20, 2000, and November 1, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: TX QUANTITY 5 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1590-1, Red Blood Cells, unit number S60865 B-1591-1, Platelets, unit number S60865 REASON: Blood products, collected from an unsuitable donor based on travel to an area considered endemic for malaria, were distributed. MANUFACTURER/RECALLING FIRM: Michigan Community Blood Center, Grand Valley Blood Program, Grand Rapids,MI RECALLED BY: The recalling firm, letter on November 29, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI QUANTITY 2 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1597-1, Red Blood Cells, Leukocytes Reduced, unit number 3824411 B-1598-1, Platelets, unit number 3824411 B-1599-1, Source Leukocytes, unit number 3824411 B-1600-1, Recovered Plasma, unit number 3824411 REASON: Blood products, collected from an unsuitable donor due to a history of residing in the United Kingdom for more than 6 months, were distributed. MANUFACTURER/RECALLING FIRM: Gulf Coast Regional Blood Center, Houston, TX RECALLED BY: The recalling firm, by letters on August 2, 2000, and April 23, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: TX, NJ and NY QUANTITY 4 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1601-1, Red Blood Cells, units 6241118, 6144105 B-1602-1, Recovered Plasma, units 6241118, 6144105 REASON: Blood products, collected from a donor with a history of having tested positive for anti-HCV, were distributed. MANUFACTURER Michigan Community Blood Center, Saginaw Valley Blood Program, Saginaw, Michigan RECALLED BY: Michigan Community Blood Center, Grand Valley Blood Program, Grand Rapids, Michigan, by letters on or about September 19, 2000 and a Swiss consignee by fax on September 21, 2001 FIRM INITIATED RECALL: Complete DISTRIBUTION: MI and Switzerland QUANTITY 4 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1606-1, Source Plasma, units 0420888166, 0420888436, 0420890211, 420890474, 0420891296, 0420891501, 0420892022, 0420892407, 0420893055, 0420893454 REASON: Blood products, collected from a donor who engaged in high-risk behavior, were distributed. MANUFACTURER North American Biologicals, Inc., Lansing, MI RECALLED BY: North American Biologicals, Inc., Boca Raton, FL, by fax dated September 19, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: NC QUANTITY 10 _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1607-1, Red Blood Cells, units 5068746, 5021652, 5006567, 4917781, 4865822, 4809574, 4750536 B-1608-1, Platelets, units 5068746, 4089574 B-1609-1, Plasma, unit 5021652 B-1610-1, Cryoprecipitate, units 5021652, 5006567 B-1611-1, Recovered Plasma, units 5068746, 4917781, 4089574, 4750536 REASON: Blood products, collected from a donor with a history of hepatitis, were distributed. MANUFACTURER/RECALLING FIRM: Carter BloodCare, Dallas, TX RECALLED BY: The recalling firm, by letters dated January 15, 1999. FIRM INITIATED RECALL: Complete DISTRIBUTION: TX QUANTITY 16 units _______ RECALL NUMBER, PRODUCT AND CODE: B-1617-1, Hormodendrum cladosporioides, lot 202032000 REASON: Allergenic extract, which failed initial bulk sterility testing, was distributed. MANUFACTURER/RECALLING FIRM: Allergy Laboratories, Inc., Oklahoma City, OK RECALLED BY: The recalling firm, by letter dated February 13, 2001. FIRM INITIATED RECALL: Ongoing DISTRIBUTION: OK QUANTITY 1 lot

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS III

_______________________________ RECALL NUMBER, PRODUCT AND CODE: B-533-1, Red Blood Cells, unit 18FT32301 REASON: Blood product, which was determined to have a hematocrit value greater than 80%, was distributed. MANUFACTURER/RECALLING FIRM: American Red Cross, Great Lakes Region, Lansing, MI RECALLED BY: The recalling firm, by telephone and letter on April 25, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1517-1, Red Blood Cells Units 38L44237, 38FE33563, 38FE33278, 38K59003, 38E70908, 38K59036, 38E70946, 38E77059, 38FE35349, 38FE33509, 38FE32769, 38L43220. B-1518-1, Red blood cells, Leukoreduced, Unit 38FC36671 B-1519-1, Platelets, Units 38FC36671, 38E70946 B-1520-1, Fresh Frozen Plasma, Units 38E70946, 38FE32769 B-1521-1, Recovered Plasma, Units 38L44550, 38FE33563, 38K59003, 38K59036,38L43220 REASON: Blood products, collected from donors who were not asked the question regarding the use of the drug Propecia, were distributed. MANUFACTURER/RECALLING FIRM: American Red Cross Blood Services, Fort Wayne, IN RECALLED BY: The recalling firm, by letters or fax dated February 16, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: CA, IN, MI, NY, OH and Switzerland QUANTITY 22 units ________________________________ RECALL NUMBER, PRODUCT AND CODE: B-1530-1, Whole Blood (1 unit), unit number 0276103 B-1531-1, Red Blood Cells (4 units), unit numbers 0249088, 2145854, 4102609, and 4112519 B-1532-1, Red Blood Cells, Leukocytes Reduced (4 units), unit numbers 0269990, 2166487, 0306783, and 2178914 B-1533-1, Platelets (3 units), unit numbers 0249088, 2145854, and 4102609 B-1534-1, Cryoprecipitated AHF (1 unit), unit number 2178914 B-1535-1, Plasma (1 unit), unit number 2178914 B-1536-1, Fresh Frozen Plasma (4 units), unit numbers 0249088, 2145854, 4102609, and 0269990 REASON: Blood products tested negative for antibodies to human immunodeficiency virus, types 1 and 2 (anti-HIV-1/2), but were collected from a donor that previously tested repeatedly reactive for anti-HIV-1, were distributed.. MANUFACTURER/RECALLING FIRM: Community Blood Center of Greater Kansas City, Kansas City, MO RECALLED BY: The recalling firm, by letter dated July 11, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: MO, KS, NE QUANTITY 18 units ________________________________ RECALL NUMBER, PRODUCT AND CODE: B-1450-1, Recovered Plasma for use in Manufacturing Injectable products, unit 1727888 REASON: Blood product, collected from a donor with a history of Hodgkin’s Disease, were distributed. MANUFACTURER/RECALLING FIRM: Central Indiana Regional Blood Center, Inc., Indianapolis, IN RECALLED BY: The recalling firm, by letters, dated September 8, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: Switzerland QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1472-1, Limulus Amebocyte Lysate (LAL) Reagent Water, lot numbers 8L5410 and 9L3690 REASON: LAL Reagent Water, that may cause inhibition of the LAL assay, was distributed. MANUFACTURER/RECALLING FIRM: BioWhittaker, Inc., Walkersville, MD RECALLED BY: The recalling firm, by telephone or facsimile on August 4, 2000 and November 1, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: Nationwide and Mexico, Brazil, South Africa, Italy, Belgium, Sweden, Canada, and Pakistan QUANTITY 2 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1505-1, Platelets, unit 01KX48677 REASON: Blood product, prepared from a unit of Whole Blood with an extended collection time, was distributed. MANUFACTURER/RECALLING FIRM: American Red Cross Blood Services, New York-Penn Region, West Henrietta, NY RECALLED BY: The recalling firm notified the consignee by letters dated December 22, 2000 and February 20, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: NY QUANTITY 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1512-1, Red Blood Cells, unit 53FL89870 REASON: Blood product, which was determined to have a hematocrit value greater than 80%, was distributed. MANUFACTURER/RECALLING FIRM: American Red Cross, Greater Chesapeake and Potomac Region, Baltimore, MD RECALLED BY: The recalling firm, by telephone on December 16, 2000 and by letter dated December 27, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: MD QUANTITY: 1 unit _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1522-1, Red Blood Cells, unit number 13FY56324 REASON: Blood product, incorrectly tested for the Kell antigen, was distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Southeastern Michigan Region, Detroit, MI RECALLED BY: The recalling firm, of the recall by letter on July 11, 2000. FIRM INITIATED RECALL: Complete DISTRIBUTION: MI QUANTITY: 1 unit. _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1546-1, Red Blood Cells, unit numbers 38FC69047 and 38FC69035 REASON: Blood products, labeled with a volume that was below the firm’s minimum specification, were distributed. MANUFACTURER/RECALLING FIRM: The American National Red Cross, Indiana-Ohio Region, Fort Wayne, IN RECALLED BY: The recalling firm, by telephone on January 16, 2001, and by letter dated February 5, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: IN QUANTITY 2 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: B-1615-1, American Elm Allergenic Extract, lot B710241A B-1616-1, White Ash Allergenic Extract, lot B712022A REASON: Allergenic extracts, which were labeled with incorrect expiration dates, were distributed. MANUFACTURER/RECALLING FIRM: Allergy Laboratories of Ohio, Inc., Columbus, OH RECALLED BY: The recalling firm, by facsimile on February 9, 2001, and by letter dated February 8, 2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: MD QUANTITY 2 lots

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

_______________________________ RECALL NUMBER, PRODUCT AND CODE: Z-327-1, X-Ray Table, XRT 600 REASON: The units were found defective under 21 CFR 1003.2, in that they fail to comply with performance standards with regards to: importation, establishing and maintaining manufacturing records and finished device records, no product report submitted and an inadequate certification program. MANUFACTURER: Imago Radiology, Abbiategrosso, Italy. RECALLED BY: Wuestec Medical, Mobile AL, telephoned the firm on June 7, 2001. FIRM INITIATED RECALL: Ongoing DISTRIBUTION: Nationwide QUANTITY 2 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: Z-0631-1 Linear MC150-C Manual Collimator, catalog #70-631xx Z-0632-1 Linear MC150 Manual Collimator, catalog #70-630xx Z-0633-1 Linear II Automatic Collimator, catalog #70-4300x Z-0634-1 Linear III Automatic Collimator, catalog #70-53xxx Z-0635-1 Linear IV Automatic Continuous SID Collimator, catalog #70-750xx REASON: Collimator housing may separate from the tube mounting plate MANUFACTURER/RECALLING FIRM: PROGENY, INC. Buffalo Grove, IL RECALLED BY: Recalling Firm, by letters dated 1/24/01. FIRM INITIATED RECALL: Ongoing DISTRIBUTION: Nationwide, Canada, Turkey QUANTITY 224 collimators ________________________________ RECALL NUMBER, PRODUCT AND CODE: Z-0636-1, Osmo 23G series reverse osmosis machine with the Permeate Divert Option. These are intended to purify the water for use in hemodialysis applications. REASON: Potential failure of Permeate valve could allow feed water into permeate stream MANUFACTURER/RECALLING FIRM: OSMONICS, INC. Minnetonka, MN RECALLED BY: OSMONICS, INC. KENT, WA, on 3/2&5/01 by letters, and by telephone on 3/6/01. FIRM INITIATED RECALL: Ongoing DISTRIBUTION: Nationwide and Canada QUANTITY 386 _______________________________ RECALL NUMBER, PRODUCT AND CODE: Z-0637-01 thru Z-0638-01 Intra-Ocular Lens. Silicone based. These lens are sterile and foldable. The affected lens come in two styles - plated haptics and a modified C loop haptic. Various models and diopters are involved. Each IOL is contained inside a small cardboard box that is labeled in part, “STAAR IOL FOLDABLE INTRAOCULAR LENS *** FOR SINGLE USE ONLY *** STERILE *** *** MODEL *** DIOPTER”. 510K/PMA PMA P910059 for models AA; and P910060 for models AQ Only model numbers with a prefix of "AA" (plate haptic) or "AQ" (loop haptic) may be affected. Serial numbers are too numerous to list as they are not consecutive. REASON: Increased tear rate of lens upon insertion into the eye. MANUFACTURER/RECALLING FIRM: STAAR SURGICAL CO., MONROVIA, CA RECALLED BY: Recalling Firm by First Class letter dated 4/5/2001. FIRM INITIATED RECALL: Complete DISTRIBUTION: Nationwide. Internationally to Canada, Newfoundland, Puerto Rico, Nova Scotia. QUANTITY 25,000 lenses

RECALLS AND FIELD CORRECTIONS: VETMED -- CLASS II

_______________________________ RECALL NUMBER, PRODUCT AND CODE: V-351-1, Veterinary Device - Anesco brand Mapleson C breathing circuits, Ref #A1003. REASON: Potential occlusion due to excess plastic in the fresh gas inlet of the elbow MANUFACTURER/RECALLING FIRM: King Systems Corportion, Noblesville, Indiana RECALLED BY: Recalling Firm, by letters on January 30, 2001. FIRM INITIATED RECALL: Ongoing. DISTRIBUTION: Nationwide, Canada, Japan and Mexico QUANTITY: 21,680 units _______________________________ RECALL NUMBER, PRODUCT AND CODE: V-352-1, Bulk Lamb Meal, All lots of bulk lamb meal shipped by the recalling firm REASON: The product is not labeled with the required caution statement “Do not feed to Cattle or other Ruminants.” MANUFACTURER/RECALLING FIRM: International Proteins Corporations (IPC), St. Paul MN RECALLED BY: Recalling Firm, Revised labeling by letter on April 17, 2001. FIRM INITIATED RECALL: Ongoing. DISTRIBUTION: MN, IL, MO, AR and TX QUANTITY 3,094 tons END OF ENFORCEMENT REPORT FOR July 04, 2001.
####

FDA News Page   |   FDA Home Page


Office of Public Affairs
Hypertext uploaded by smc 2001-JUL-03.