March 28, 2001                                               01-12


RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II

________ PRODUCT: Source Plasma. Recall #B-833-1. CODE: Units 0020758752, 0020759244, 0020759466, 0020759868, 0020760080. MANUFACTURER: North American Biologicals, Inc., Norfolk, VA. RECALLED BY: Manufacturer, by facsimile on July 25, 2000. Firm-initiated recall complete. DISTRIBUTION: North Carolina. QUANTITY: Five units. REASON: Blood products, collected from a donor who had not completed the CJD/nvCJD increased risk questions, were distributed. ________ PRODUCT: Red Blood Cells. Recall #B-834-1. CODE: Units 5507064, 5508438, 5509671. MANUFACTURER: LifeSource, Glenview, IL . RECALLED BY: Manufacturer, by telephone on May 24, 2000, and July 25, 2000. Firm-initiated recall complete. DISTRIBUTION: Illinois. QUANTITY: Three units. REASON: Blood products, collected from a donor who was taking Tamoxifen, were distributed. ________ PRODUCT: a) Red Blood Cells. Recall #B-931-1; b) Platelets. Recall #B-932-1. CODE: a) and b) unit number 13FS86009. MANUFACTURER: The American National Red Cross, Southeastern Michigan Region, Detroit, MI. RECALLED BY: Manufacturer, by letter dated December 19, 2000. Firm-initiated recall complete. DISTRIBUTION: Michigan. QUANTITY: Two units. REASON: Blood products, collected from an unsuitable donor based on travel to an area considered endemic for malaria, were distributed. ________ PRODUCT: Red Blood Cells. Recall # B-933-1. CODE: Unit #53H77408. MANUFACTURER: The American National Red Cross, Greater Chesapeake and Potomac Region, Baltimore, MD. RECALLED BY: Manufacturer, by letter dated November 13, 2000. Firm-initiated recall complete. DISTRIBUTION: Maryland. QUANTITY: One unit. REASON: Blood product, collected from an unsuitable donor based on travel to an area considered endemic for malaria, was distributed. ________ PRODUCT: Platelets Pheresis (210 units). Recall #B-935-1. CODE: Unit numbers 18P43116, 18P43117, 18P43122, 18P43124, 18P43128, 18P43138, 18P43166, 18P43167, 18P43170, 18P43171, 18P43172, 18P43173, 18P43174, 18P43175, 18P43176, 18P43177, 18P43178, 18P43179, 18P43180, 18P43182, 18P43183, 18P43184, 18P43185, 18P43186, 18P43187, 18P43188, 18P43189, 18P43191, 18P43192, 18P43193, 18P43196, 18P43199, 18P43202, 18P43204, 18P43209, 18P43210, 18P43211, 18P43212, 18P43213, 18P43214, 18P43215, 18P43216, 18P43217, 18P43218, 18P43219, 18P43220, 18P43221, 18P43222, 18P43224, 18P43225, 18P43227, 18P43229, 18P43230, 18P43232, 18P43233, 18P43234, 18P43235, 18P43237, 18P43238, 18P43239, 18P43242, 18P43243, 18P43244, 18P43245, 18P43247, 18P43248, 18P43249, 18P43250, 18P43251, 18P43252, 18P43253, 18P43254, 18P43256, 18P43258, 18P43259, 18P43260, 18P46262, 18P43263, 18P43264, 18P43265, 18P43267, 18P43269, 18P43270, 18P43271, 18P43273, 18P43276, 18P43277, 18P43278, 18P43279, 18P43280, 18P43281, 18P43282, 18P43283, 18P43284, 18P43286, 18P43287, 18P43288, 18P43289, 18P43290, 18P43291, 18P43293, 18P43294, 18P43295, 18P43297, 18P43298, 18P43299, 18P43300, 18P43301, 18P43302, 18P43303, 18P43304, 18P43305, 18P43306, 18P43308, 18P43311, 18P43312, 18P43314, 18P43316, 18P43317, 18P43318, 18P43323, 18P43324, 18P43325, 18P43326, 18P43327, 18P43328, 18P43329, 18P43331, 18P43332, 18P43333, 18P43334, 18P43335, 18P43336, 18P43337, 18P43338, 18P43339, 18P43340, 18P43341, 18P43343, 18P43344, 18P43345, 18P43347, 18P43348, 18P43349, 18P43350, 18P43351, 18P43352, 18P43353, 18P43354, 18P43355, 18P43356, 18P43358, 18P43359, 18P43360, 18P43361, 18P43362, 18P43363, 18P43364, 18P43365, 18P43366, 18P43367, 18P43368, 18P43369, 18P43370, 18P43371, 18P43372, 18P43373, 18P43374, 18P43376, 18P43377, 18P43379, 18P43380, 18P43382, 18P43383, 18P43385, 18P43386, 18P43387, 18P43388, 18P43389, 18P43390, 18P43391, 18P43392, 18P43393, 18P43394, 18P43395, 18P43396, 18P43397, 18P43399, 18P43400, 18P43401, 18P43403, 18P43404, 18P43407, 18P43410, 18P43411, 18P43412, 18P43413, 18P43414, 18P43415, 18P43416, 18P43114, 18P43115, 18P43118, 18P43119, 18P43120, 18P43121, 18P43123, 18P43125, 18P43126 and 18P43417. MANUFACTURER: The American National Red Cross, Great Lakes Region, Lansing, MI. RECALLED BY: Manufacturer, by telephone beginning on July 18, 2000, and by letter beginning August 1, 2000. Firm-initiated recall complete. DISTRIBUTION: Michigan. QUANTITY: 210 units. REASON: Blood products, collected in bags that have loose or missing tabs that cover the sterile ports, were distributed.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS III

________ PRODUCT: Recovered Plasma for Manufacturing Injectables. Recall #B-835-1. CODE: Units 5507064, 5508438, 5509671. MANUFACTURER: LifeSource, Glenview, IL. RECALLED BY: Manufacturer by facsimile on May 26, 2000 and July 28, 2000. Firm- initiated recall complete. DISTRIBUTION: California. QUANTITY: Three units. REASON: Blood products, collected from a donor who was taking Tamoxifen, were distributed. ________ PRODUCT: Source Plasma. Recall #B-836-1. CODE: Units 135563, 139323, 139683, 139914. MANUFACTURER: Sera Tec Biologicals Limited Partnership, Erie, PA. RECALLED BY: Sera Tec Biologicals Limited Partnership, North Brunswick, NJ., by facsimile on December 27, 2000. Firm-initiated recall complete. DISTRIBUTION: North Carolina. QUANTITY: Four units. REASON: Blood products, collected from donors who subsequently tested repeat reactive for anti HCV, were distributed. __________ PRODUCTS: a) Plasma, Cryoprecipitate Reduced (12 units). Recall #B-929-1 b) Fresh Frozen Plasma (61 units). Recall #B-930-1. CODES: a) unit numbers 18FJ31448, 18FJ31467, 18FQ55714, 18FQ55722, 18FS37894, 18FS38091, 18FS40822, 18FT30485, 18FV15617, 18FZ00067, 18FZ00101, and 18FZ02434; b) unit numbers 18FW08325, 18FJ32984, 18FK40622, 18FS42226, 18FS42260, 18FT32276, 18FV20125, 18FV20130, 18FV20149, 18FX15830, 18FF05994, 18FV19880, 18FV19887, 18FV19895, 18FV19902, 18FZ04865, 18FZ04871, 18FZ04883, 18FZ04899, 18FQ57057, 18FT32385, 18FZ04764, 18FZ04829, 18FE69023, 18FT32270, 18FE68878, 18FE68885, 18FE68887, 18FV19865, 18FV19867, 18FV19875, 18FQ56791, 18FQ56793, 18FQ56795, 18FQ56796, 18FQ56803, 18FQ56810, 18FQ56817, 18FT31936, 18FV19627, 18FZ04669, 18FZ04672, 18FZ04678, 18FR47263, 18FT32241, 18FW08261, 18N72922, 18FV19913, 18FR47290, 18FR47344, 18FV19884, 18FV19911, 18FV19912, 18FX15869, 18FX15876, 18GG32936, 18FF05989, 18FZ04860, 18FK40083, 18FK40111 and 18FQ56568 MANUFACTURER: The American National Red Cross, Great Lakes Region, Lansing, MI. RECALLED BY: Manufacturer, by telephone on April 12, 2000, and by letter dated April 13, 2000. Firm-initiated recall complete. DISTRIBUTION: Michigan. QUANTITY: a) 12units; b)61 units. REASON: Blood products, subjected to elevated temperatures during storage, were distributed.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

________ PRODUCT: Rugged EZ-Pro 2, Model No. 6091 Ambulance Cots. Recall Number Z-187-1. CODE: Serial Nos. 990839435 through 001139084. MANUFACTURER: Stryker Medical, Kalamazoo, MI. RECALLED BY: Manufacturer, by notification letters to consignees between December 1 and December 5, 2000 advising them of upcoming field corrections by Stryker service technicians. Firm-initiated field corrections ongoing. DISTRIBUTION: Nationwide and Canada, Japan, Spain, Switzerland, United Kingdom, and Mexico. QUANTITY: 919 total units. REASON: The device lock pin may not properly engage and therefore the cot may drop to the next lower level while transporting patients. NOTE: This is an extension of the firm's original field correction of August 21, 2000. This action has been extended to include additional serial numbers and provides an additional correction for the problem. ________ PRODUCT: Saline, 0.45%, 1.8 ml., Product No. D12, intended for use in preparation of standardized suspensions of microorganisms for use in Vitek automated identification and susceptibility test system. Recall No. Z-321-1. CODE: Lot Nos. 0350, 0357, 1005, and 1012. MANUFACTURER: Hardy Diagnostics, Santa Maria, CA. RECALLED BY: Manufacturer, by telephone followed by letters on February 14, 2001. Firm-initiated recall ongoing. DISTRIBUTION: CA, MT, PA, NV, WY and Puerto Rico. QUANTITY: 23,750 units. REASON: Use of this product on certain Vitek susceptibility panels will result in insufficient growth of gram negative microbes.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III

________ PRODUCT: STA LiaTest D-DI Kit, Model No. 0515. Recall #Z-280-1. CODE: Lot No. 981872, Exp. 1/2000. MANUFACTURER: Diagnostica Stago, Inc., Parsippancy, NJ. RECALLED BY: Manufacturer, by contact on/about August 6, 1999. Firm-initiated recall complete. DISTRIBUTION: Colorado and Indiana. QUANTITY: 40units. REASON: The product non-conformity is that the recovery of abnormal control is too low, in the order of 15 to 20% compared to the acceptable value. _______ PRODUCT: Cobas Mira Creatinine Reagent: a) Catalog No. 3033414. Recall #Z-313-1; b) Catalog No. 3034712. Recall #Z-314-1. CODE: a) Lot No. A12071 Exp. Date 9/30/2002, Lot No. A12076 Exp. Date 2/28/2003; b) Lot No. A12075 Exp. Date 10/31/2002, Lot No. A12077 Exp. Date 2/28/00. MANUFACTURER: Roche Diagnostics Corp., Indianapolis, IN RECALLED BY: Manufacturer, by letter on December 20,2000. Firm-initiated recall ongoing. DISTRIBUTION: Nationwide. QUANTITY: 4,112 kits. REASON: The device may produce falsely elevated creatinine results due to increased triglycerides in serum samples. ________ PRODUCT: ROCHE c.f.a.s. Calibrator for Automated Systems, Catalog No. 759350. Recall #Z-315-1. CODE: Lot Nos. 199608 through 199611, 199845 through 199848. MANUFACTURER: Roche Diagnostics Corp., Indianapolis, IN. RECALLED BY: Manufacturer, by Notice on December 26, 2000. Firm-initiated recall complete. DISTRIBUTION: United States. QUANTITY: Approximately 10,000 kits. REASON: A slight positive bias of the product setpoints causes serum calcium results too high. ________ PRODUCT: Roche Elecsys 2010 assay tips, Catalog No. 1706799. Recall #Z-318-1. CODE: All Lots beginning with the numbers 051. MANUFACTURER: Roche Diagnostics Corp., Indianapolis, IN. RECALLED BY: Manufacturer, by Notice on January 22, 2001. Firm-initiated recall ongoing. DISTRIBUTION: United States. QUANTITY: Approximately 10,000 kits. REASON: Some tips used to pipette samples and reagents may have plugged openings that may result in incorrect results. ________ PRODUCT: Access Immunoassay System; Troponin I Reagent Pack. Recall #Z-319-1. CODE: Lot Nos. 093856 and 093859. MANUFACTURER: Beckman Coulter Inc., Brea, CA. RECALLED BY: Manufacturer, by letter on February 7, 2001. Firm-initiated recall ongoing. DISTRIBUTION: Nationwide and International. QUANTITY: 1,482 units and 2,352 units. REASON: Some Troponin Reagent packs may not contain sufficient reagent to complete the 50 tests indicated in the labeling. ________ PRODUCT: STA LiaTest von Williebrand vWF Factor Antigen Kits, Catalog No. 00518. Recall #Z-320-1. CODE: Lot No. 982721, Exp. 3/21/2000. MANUFACTURER: Diagnostica Stago, Inc., Parsippancy, NJ. RECALLED BY: Manufacturer, by letter on January 14, 2000. Firm-initiated recall complete. DISTRIBUTION: Nationwide and International. QUANTITY: 165 kits. REASON: We are assigning the contamination of the product was due to the Lab Technician inappropriately handling the product during the quality control analysis.

RECALLS AND FIELD CORRECTIONS: VETERINARY MED -- CLASS II

________ PRODUCT: Iams brand Large Breed Adult Formula Dog Food, packed in 8 lb. and 40 lb. Bags. Recall #V-012-1. CODE: Best Used By Date: "AUG 20 01"/Code Date: "200400 PB4133". MANUFACTURER: The Iams Company, Dayton, Ohio. RECALLED BY: Manufacturer, via telephone and memorandum on June 6, 2000. Update letters, dated June 8, June 9, June 21 and June 26, 2000, were faxed and mailed to distributors and sales Representatives who visited retail locations. Firm- initiated recall complete. DISTRIBUTION: New York, Pennsylvania, Rhode Island, Michigan, Ohio, Indiana, and Kentucky. QUANTITY: 248,080 lbs. REASON: Product contains high levels of DL-Methionine Amino Acid.

RECALLS AND FIELD CORRECTIONS: VETERINARY MED -- CLASS III

________ PRODUCT: Erythromycin Phosphate, Non-Sterile, For Veterinary Use Only - A bulk drug substance intended for the manufacture of various veterinary antibiotics for the treatment of respiratory infections in cattle, swine and sheep and for the treatment of fish ick. Product Code: 27919, List Numbers 27919-29-06, 05151-04-05, and 05151-04-11. Recall #V-013-1. CODES: Lots: 67840CD00, 68956CD00, 68975CD00, 68976CD00, 68982CD00, 69034CD00, and 69035CD00. MANUFACTURER: Abbott Laboratories, Inc., Specialty Products Division, North Chicago, Illinois. RECALLED BY: Manufacturer, by telephone on January 19, and January 26, 2001 and letters dated January 22, and January 30, 2001. Firm-initiated recall ongoing. DISTRIBUTION: California, Minnesota, Washington, Texas, and Connecticut. France and Canada. QUANTITY: 2,396.11 KA. REASON: The bulk drug substance was contaminated with trace amounts (35 ppm) of ethylene glycol.
END OF ENFORCEMENT REPORT FOR MARCH 28 2001.
####