The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

March 4, 1998                                         98-09

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS I =============
_______________
PRODUCT        Ice cream, various flavors labeled for Food Lion,
               packaged in ´ gallon cartons and 5 quart plastic
               pails:
               a) Food Lion Vanilla Flavored Ice Cream, ´ gallon
               and 5 quart sizes coded 7-13-98
               b) Food Lion Neapolitan Flavored Ice Cream, 5          quart size coded 7-13-98 
               c) Food Lion Fudge Royale Ice Cream, 5 quart size 
               coded 7-14-98 
               d) Food Lion Chocolate Ice Cream, 5 quart size.  
               Recall #F-273/276-8.
CODE           a) 7-13-98; b) 7-13-98; c) 7-14-98; d) 7-14-98.
MANUFACTURER   Dairy Fresh, Inc., Winston-Salem, North Carolina.
RECALLED BY    Manufacturer, by press release followed by
               telephone on October 27, 1997.  Completed recall
               resulted from sample analysis and follow-up by the
               North Carolina Department of Agriculture, Food and
               Drug Division.
DISTRIBUTION   Delaware, Florida, Georgia, Kentucky, Maryland,
               North Carolina, Pennsylvania, South Carolina,
               Tennessee, Virginia, West Virginia.
QUANTITY       131,000 containers were distributed; firm
               estimates none remains on the market.
REASON         Products may have been contaminated with Listeria
               monocytogenes.


RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS III ===========
_______________
PRODUCT        Mayonnaise and salad dressing packaged in 48 fluid
               ounce plastic jars:
               a) Kraft Mayo Real Mayonnaise, UPC #21000-64014 
               b) Kraft Miracle Whip Salad Dressing, UPC
               #21000-64514 
               c) Kraft Miracle Whip Light Dressing, UPC
               #21000-64595.  Recall #F-270/272-8.
CODE           a) JUL 13 98 D, JUL 14 98 D, AUG 13 98 D,                   AUG 14 98 D
               b) JUL 15 98 D, JUL 16 98 D, JUL 17 98 D, AUG 10
               98 D, AUG 11 98 D, AUG 12 98 D,  JUN 23 98 D and
               JUN 24 98 D
               c) JUL 14 98 D, JUL 15 98 D, AUG 13 98 D, AUG 25
               98 D.
MANUFACTURER   Kraft Foods, Inc., Garland, Texas.
RECALLED BY    Kraft Foods, Inc., Northfield, Illinois, by
               memorandum dated February 9, 1998 and by fax or
               visit on or about February 16, 1998.         
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       42,339 cases (6 jars per case) were distributed.
REASON         Products are in containers which are short weight      by 2 fluid ounces, 


RECALLS AND FIELD CORRECTIONS: DRUGS -- CLASS II
============
_______________
PRODUCT        Slow Fe Tablets, (160 mg dried ferrous sulfate,
               USP, equiv. to 50 mg elemental iron), OTC, blister
               packaged units of 90 units, indicated for the
               treatment of anemia.  
               Recall #D-079-8.
CODE           Lot numbers: U09868K EXP 7/99, U10076K EXP 8/99,            U10077K EXP 8/99.
MANUFACTURER   Novartis Pharmaceutical, West Sussex, England.
RECALLED BY    Novartis Consumer Health, Inc., Summit, New
               Jersey, by telephone on September 4, 1997, and
               November 24, 1997, and by letters mailed on
               September 5, 1997, and December 12, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Georgia, Pennsylvania, Nebraska.
QUANTITY       39,324 cartons were distributed.
REASON         Labeling lacks tetracycline, nursing, and
               pregnancy warnings.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Upjohn brand Cortef Oral Suspension
               (Hydrocortisone Cypionate oral suspension, USP),
               10 mg per 5 mL, in 4 fluid ounce bottles, Rx oral anti-inflammatory.  
               NDC #0009-0142-01.  Recall #D-080-8.
CODE           Lot #86BTF EXP 10/31/02.
MANUFACTURER   Pharmacia & Upjohn, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by telephone on January 27, 1998. 
               Firm-initiated recall complete.
DISTRIBUTION   Connecticut, Georgia, Indiana, Massachusetts, New
               Jersey, New York, North Carolina, Ohio, Virginia,
               The United Kingdom.
QUANTITY       108 bottles were distributed.
REASON         Subpotent.

_______________
PRODUCT        Glyburide Tablets, 2.5 mg, in 90 tablet bottles,
               Rx oral blood-glucose-lowering drug of the
               sulfonylurea class.  NDC #59762-3726-6.  Recall
               #D-081-8.
CODE           Lot #135JU EXP 7/98.
MANUFACTURER   Pharmacia & Upjohn, Puerto Rico.
RECALLED BY    Greenstone Limited (a subsidiary of Pharmacia &
               Upjohn Co.) Portage, Michigan, by letter on
               January 5, 1998.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       10,426 bottles were distributed.
REASON         Mislabeling - Product bears the incorrect
               expiration date.

_______________
PRODUCT        Prozac Pulvules (Fluoxetine Hydrochloride    capsules), 10 mg, in units of 100 and 2000 count  bottles, Rx indicated for the treatment of
               depression, obsessive-compulsive disorder and
               bulimia nervosa, under the Dista label.
               NDC #0777-3104-02 BOTTLES OF 100
               NDC  0777-3104-07 BOTTLES OF 2000.
               Recall #D-082-8. 
CODE           LOT NUMBER     EXP. DATE      PACKAGE
               1AH81B         JULY 1, 1999   100 CAPSULES
               1AH82B         JULY 1, 1999   100     "
               1AH83M         JULY 1, 1999   2000    "
               1AG33M         JULY 1, 1999   2000    "
               1AG34M         JULY 1, 1999   2000    "
               1AG35M         JULY 1, 1999   2000    ".
MANUFACTURER   Lilly Del Caribe, Inc., Carolina, Puerto Rico.
               Shionogi Qualicaps, Whitsett, North Carolina
               (capsule supplier).
RECALLED BY    Eli Lilly Company, Indianapolis, Indiana, by
               letter and by fax on January 22, 1998.        
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       106,190 bottles were distributed.
REASON         Misbranded - Some capsules are incorrectly
               imprinted as "20 mg units" and incorrectly item
               coded for 20 mg capsules as "3105".

_______________
PRODUCT        Prempro Tablets (Conjugated Estrogens 
               0.625mg/Medroxyprogesterone Acetate 2.5mg ), in
               unit dose cartons of 28 tablets (14 tablets per
               card) sold in cartons of 3, prescription hormone
               replacement therapy.
               NDC #0046-0875-01/02.  Recall #D-083-8.
CODE           Lot #9970927 EXP 7/99. 
MANUFACTURER   Wyeth-Ayerst Laboratories, Rouses Point, New York.
RECALLED BY    Wyeth Ayerst Laboratories/Whitehall Robbins,
               Richmond, Virginia, by letter dated January 29,
               1998.  Firm-initiated recall ongoing.
DISTRIBUTION   Mid-Western United States.
QUANTITY       24,931 cartons were distributed.
REASON         Content uniformity failure (medroxyprogesterone
               acetate superpotent).

_______________
PRODUCT        Xtra-Lax Laxative Tablets (Phenolphthalein), in
               boxes of 30, 60, and 75 tablets, OTC product for
               the relief of irregularity, under the following
               brand names:  Guardian Drug Co. (GDC), Kenney,
               Mason Remedies, Drug Imporium, Q-Lax          
               (NDC 0603-0260-16), Relief Plus. 
Recall         #D-084-8.
CODE           LOT #               EXP. DATE
               142-3071               01/98
               142-3111               04/98
               142-3237               05/98
               142-3469               10/98
               142-3565               11/98
               142-3756               04/99
               142-3828               05/99.
MANUFACTURER   Guardian Drug Company, Trenton, New Jersey.
RECALLED BY    Manufacturer, by letter on October 24, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 2,538 cases were distributed.
REASON         Product failed disintegration testing.


RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II ========
_______________
PRODUCT        a) Red Blood Cells; b) Platelets; c) Recovered
               Plasma.  Recall #B-542/544-8.
CODE           Unit #53H09225.
MANUFACTURER   American Red Cross Blood Services, Baltimore,
               Maryland.
RECALLED BY    Manufacturer, by letter dated November 17, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Maryland and California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products, which tested negative for the
               antibody to the human immunodeficiency virus type 1 (anti-HIV-1), but were collected from a donor
               who previously tested repeatedly reactive for
               anti-HIV-1, Western blot negative.

_______________
PRODUCT        Albumin (Human) 25% in 100ml Vials.  
               Recall #B-568-8.
CODE           Lot Numbers: 684R004, 6C4R016A, 684R095.
MANUFACTURER   Bayer Corporation, Clayton, North Carolina.
RECALLED BY    Bayer Corporation, Elkhart, Indiana, by letter
               dated January 9, 1998.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Arizona, California, Florida, Indiana, New York,
               Pennsylvania, Tennessee, Indonesia, Bulgaria,
               Canada.
QUANTITY       15,777 vials were distributed; firm estimated that
               very little product remained on market at time of recall initiation.
REASON         Albumin, for which sterility assurance was not     
               sufficiently documented.

_______________
PRODUCT        a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-582/583-8.
CODE           Unit #53KN35680.
MANUFACTURER   American Red Cross, Baltimore, Maryland.
RECALLED BY    Manufacturer, by letter dated October 27, 1997,
               and by fax on October 24, 1997.              
               Firm-initiated recall complete.
DISTRIBUTION   District of Columbia and California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products tested negative for antibody to the
               human immunodeficiency virus type 1 (anti-HIV-1), but collected from a donor  who previously tested
               repeat reactive for anti-HIV-1, Western Blot
               negative and was incorrectly reentered.

_______________
PRODUCT        Red Blood Cells.  Recall #B-584-8.
CODE           Unit #V39366.
MANUFACTURER   Aurora Area Blood Bank, Aurora, Illinois.
RECALLED BY    Manufacturer, by letter dated December 10, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit was distributed.
REASON         Blood product tested negative for antibody to the human immunodeficiency virus type 1 (anti-HIV-1), 
               but was collected from a donor who previously
               tested repeat reactive for anti-HIV-1, Western
               Blot negative and was incorrectly reentered.

_______________
PRODUCT        a) Red Blood Cells; b) Platelets.
               Recall #B-593/594-8.
CODE           Unit #49L83598.
MANUFACTURER   American Red Cross Blood Services, Tulsa,
               Oklahoma.
RECALLED BY    Manufacturer, by letter dated November 17, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Texas and California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products were collected from a donor who
               reported travel to an area designated as endemic
               for malaria.

_______________
PRODUCT        Red Blood Cells.  Recall #B-595-8.
CODE           Unit #4918320.
MANUFACTURER   BloodCare, Dallas, Texas.
RECALLED BY    Manufacturer, by letter dated September 29, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit was distributed.
REASON         Blood products were tested negative for the
               antibody to Hepatitis B core antigen (anti-HBc),
               but was collected from a donor who previously
               tested repeat reactive on two previous occasions
               for anti-HBc.

_______________
PRODUCT        a) Red Blood Cells; b) Recovered Plasma.
Recall         #B-597/598-8.
CODE           Unit #0819167.
MANUFACTURER   Coffee Memorial Blood Center, Inc., Amarillo,
               Texas.
RECALLED BY    Manufacturer, by telephone on October 14, 1997,
               and by letter dated January 16, 1998.  Firm-initiated recall complete.
DISTRIBUTION   Tennessee and California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products tested negative for Cytomegalovirus
               (CMV) by EIA, but collected from a donor who
               previously tested repeat reactive for CMV.

_______________
PRODUCT        Platelets.  Recall #B-605-8.
CODE Unit      #03G00006.
MANUFACTURER   American Red Cross Blood Services, Atlanta,
               Georgia.
RECALLED BY    Manufacturer, by letter dated February 9, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Florida.
QUANTITY       1 unit was distributed.
REASON         Blood product was collected from a donor who
               traveled to an area considered endemic for
               malaria.

_______________
PRODUCT        a) Red Blood Cells; b) Platelets.
               Recall #B-606/607-8.
CODE           Unit #03GF25194.
MANUFACTURER   American Red Cross Blood Services, Atlanta,
               Georgia.
RECALLED BY    Manufacturer, by letter dated August 17, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Georgia.
QUANTITY       1 unit was distributed.
REASON         Blood products were collected from a donor who
               traveled to an area considered endemic for
               malaria.

_______________
PRODUCT        Red Blood Cells.  Recall #B-609-8.
CODE           Unit #21GG85881.
MANUFACTURER   American Red Cross Blood Services, Portland,
               Oregon.
RECALLED BY    Manufacturer, by letter dated January 7, 1998. 
               Firm-initiated recall complete.
DISTRIBUTION   Oregon.
QUANTITY       1 unit was distributed.
REASON         Blood product was collected from a donor who was
               exposed to hepatitis C.

_______________
PRODUCT        a) Red Blood Cells; b) Cryoprecipitated AHF.
               Recall #B-610/611-8.
CODE           Unit #21GM76062.
MANUFACTURER   American Red Cross Blood Services, Portland,
               Oregon.
RECALLED BY    Manufacturer, by telephone on December 3, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products were collected from a donor with a history of cancer.

_______________
PRODUCT        Platelets.  Recall #B-622-8.
CODE           Unit #4540677.
MANUFACTURER   Central Indiana Regional Blood Center, Inc.,
               Indianapolis, Indiana.
RECALLED BY    Manufacturer, by letter on December 1, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Indiana.
QUANTITY       1 unit was distributed.
REASON         Blood product corresponded to a unit of clotted
               Red Blood cells.

_______________
PRODUCT        a) Red Blood Cells; b) Platelets; c) Recovered    Plasma; d) Buffy Coat.  Recall #B-625/628-8.
CODE           Unit #18S63459.
MANUFACTURER   American Red Cross, Lansing, Michigan.
RECALLED BY    Manufacturer, by telephone on May 28, 1997 and by
               letters dated May 29, 1997 and April 16, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Michigan, California, New Jersey.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products were collected from two donors with
               the same whole blood number and one donor deferred
               for body piercing within twelve months of
               donation.

_______________
PRODUCT        a) Platelets; b) Fresh Frozen Plasma.
               Recall #B-631/632-8.
CODE           Unit #18FS12068.
MANUFACTURER   American Red Cross, Lansing, Michigan.
RECALLED BY    Manufacturer, by letter on October 23, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY       1 unit of each component was distributed.    
REASON         Blood products corresponded to a unit of clotted
               Red Blood Cells.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT        Platelets, Pheresis.  Recall #B-317-8.
CODE           Unit numbers:  17FX05851-12050, 17FX05851-12051.
MANUFACTURER   American Red Cross, North Central Blood Services,
               St. Paul, Minnesota.
RECALLED BY    Manufacturer, by telephone on October 1, 1996, and
               by letter dated October 11, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   Minnesota.
QUANTITY       2 units were distributed.
REASON         The product exceeded the acceptable standard for a
               leukoreduced blood component.


RECALLS AND FIELD CORRECTIONS:  DEVICES --
==================
_______________
PRODUCT        Coulter Maxm Series Analyzers, an automated
               differential cell counter:
               a) Model No. 0041, COULTER MAXM System;
               b) Model No. 0056, COULTER MAXM with AUTOLOADER;
               c) Model No. 0141, COULTER MAXM with Reticulocyte  
               Analysis;
               d) Model No. 0156, COULTER MAXM with AUTOLOADER
               and Reticulocyte Analysis;
               e) Model No. 0940, COULTER MAXM with Japanese DMS;
               f) Model No. 0941, COULTER MAXM with AUTOLOADER
               and Japanese DMS.
               Recall #Z-335/340-8.
CODE           All serial numbers.
MANUFACTURER   Coulter Corporation, Miami, Florida.
RECALLED BY    Manufacturer, by letter on August 8, 1997.    
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 2,075 units were distributed.
REASON         Components in the reagent level sensor area of the
               instrument can ignite. 

_______________
PRODUCT        Sterile Cotton Tipped Applicators (6" wood Stick) used in various custom procedure trays:
               a) Heart Pack, Product No. CV4482F
               b) Laparoscopy Pack, Product No. OBC595B
               c) Laser Laparoscopy Tray, Product No. OB4283D
               d) Basic Ophthalmology Pack, Product No. EYC024
               e) Prep Set, Product No. PRC022A
               f) Bone Tray w/ aux., Product Nos. GS6226F & G
               g) C-Section Pack, Product No. OB5170F
               h) Tubal Pack, Product No. OB5325E
               I) Open Heart Pack, Product No. CVC378B
               j) Eye Pack, Product Nos. EYL351 & A
               k) Total Joint Tray, Product No. OP6116F
               l) Pelvic Exam Tray, Product No. OB8788A
               m) Lap Chole Tray, Product Nos. LZ6324F & G,
               Product No. LZB773A, Product No. LZC594A, 
               n) Nasal Pack, Product No. NA4911E
               0) Pediatric CVP/TPN Tray, Product No. DR5088
               p) Bone Marrow Tray, Product No. GSD492B.
               Recall #Z-341/356-8.
CODE           Lot numbers:  a) Nos. 46081720, 46082617,
               46102297; b) 46122312, 47012250;
               c) 47022362; d) 46082040; e) 46111764;
               f) 46081988, 46102375, 46112056, 45112505,
               46120319,   46122560, 47010303, 47011275,
               47030845, 47031772;
               g) 46091390, 46111016, 46121656, 47010809,
               97010028RS, 47032329;
               h) 46101185, 46101340, 46120067, 47013059,
               47021050, 47051560; I) 46090748;
               j) 46081056, 46111888, 47021884;
               k) 47012922, 47022141, 47022339; l) 46082746;
               m)Product Nos. LZ6324F & G, Lot Nos.46090754,
               46100805, 46101898, Product No. LZB773A, Lot No.
               46080940, Product No. LZC594A, Lot Nos. 46082136,
               46112368, 47021591; n) Lot No. 46091335;
               o) 46082038, 46082610, 46110300, 46122626,
               47021475; p) 46082667 46090026, 46120655,
               47011568.
MANUFACTURER   Maxxim Medical, Richmond, Virginia (custom tray);
               Banta Healthcare Products, Inc. (Formerly Ling
               Products, Inc.), Neenah, Wisconsin (cotton tip
               applicator).
RECALLED BY    Maxxim Medical, Richmond, Virginia, by letter
               dated June 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       3,706 trays were distributed.
REASON         The cotton tip of the applicator may detach during
               use.

_______________
PRODUCT        EIE/Analytic Power Cart, Part #971-0100, used in
               dental offices to supply power to and serve as a
               base for other dental devices and accessories. 
               Recall #Z-360-8.
CODE           Serial numbers: 05998&9, and 060000 through
               060021.  Additionally, 6 units had no serial
               numbers assigned by ASI Medical, Inc.
MANUFACTURER   ASI, Medical, Inc., Englewood, Colorado (contract
               manufacturer).
RECALLED BY    EIE/Analytic, Division of Ormco Corporation,
               Glendora, California, by letter dated December 5,
               1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Alabama, Arizona, California, Colorado, Florida,
               Illinois, Massachusetts, Maine, Maryland, North
               Carolina, New Jersey, New Mexico, Oklahoma, Utah,
               Korea.
QUANTITY       28 units were distributed.
REASON         The installation and design of the Power Carts
               electrical system may have been a factor in the
               malfunction of the unit.

_______________
UPDATE         Recall #Z-326-8, Exactech, Inc., recall of Opteon
               brand Femoral Stems, which appeared in the
               February 18, 1998 Enforcement Report is being
               updated as follows:
PRODUCT        Opteon brand Femoral Stems, with laser cut
               alignment marks, indicated for use in skeletally
               mature individuals undergoing primary or secondary
               surgery for proximal femoral fracture where
               prosthetic replacement is determined by the
               surgeon to be the preferred treatment.   Recall
               #Z-326-8.
CODES          Part # Serial Number Range
               102-01-01  any serial number before UHKG
               102-01-02  any serial number before UGUH
               102-01-03  any serial number before UGQM
               and including UGNT-UGPB
               102-01-04  any serial number before UGTA
               102-01-05  any serial number before UGJN.


RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III =========
_______________
PRODUCT        NeuroCare Group Camino Heyer-Schulte Hermetic II
                External CSF Drainage Management System, Sterile  for Single Use Only, used to provide a sterile    fluid path for draining cerebrospinal fluid from  the ventricles of the brain or the lumbar    subarachnoid space to a collection bag.  The  primary indication for the product is external drainage in the management of hydrocephalic shunt
               infections.
               Recall #Z-334-8.
CODE           Lots 1961716, 1970452, 1970745, and 1970746.
MANUFACTURER   Heyer-Schulte NeuroCare, L.P., Anasco, Puerto
               Rico.
RECALLED BY    Heyer Schulte NeuroCare, L.P., Pleasant Prairie,
               Wisconsin, by telephone on February 5, 1998,
               followed by letter dated February 2, 1998.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       677 units were distributed.
REASON         Part of the device, a burette chamber, may break  off at the junction with the tubing and stopcock.


______________
PRODUCT        Recall #Z-329/330-8, Maxima Forte Oxygenator      (Medtronic, Inc., Anaheim, California), which     appeared in the February 25, 1998 Enforcement     Report is a Class III Recall.



END OF ENFORCEMENT REPORT FOR MARCH 4, 1998.  BLANK PAGES MAY FOLLOW.

                                   ####