The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
07/27/1994

Recalls and Field Corrections:  Foods -- Class I -- 07/27/1994

July 27, 1994                                                     94-30
               
RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        C. Ochoa White Cheese in 8 and 12 ounce vacuum packed
               plastic packages.  Recall #F-737-4.
CODE           One uncoded lot manufactured March 29, 1994.
MANUFACTURER   C. Ochoa Cheese, Brownsville, Texas.
RECALLED BY    Manufacturer, by visit April 4 and 5, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas.
QUANTITY       Firm estimates none remains on the market.
REASON         Product had the potential to support the outgrowth of
               Clostridium botulinum toxin.

_______________
PRODUCT        Pepperidge Farm Goldfish Snack Mix Nutty Deluxe with Honey
               Roasted Cashews/Almonds, in 9-1/4 ounce cartons.
               Recall #F-738-4.
CODE           All pull dates through Sell By 10/07/94.
MANUFACTURER   Pepperidge Farm, Inc., Willard, Ohio.
RECALLED BY    Pepperidge Farm, Inc., Norwalk, Connecticut, by press
               release May 27, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and the Caribbean.
QUANTITY       Approximately 14,000 to 15,000 cases (24 boxes per case).
REASON         Product contains nuts which were roasted in cold pressed
               peanut oil and the label fails to declare the peanut oil
               ingredient.
_______________
PRODUCT        Hot Smoked Catfish Fillets, random weights, in either vacuum
               packed plastic bags, vacuum packed plastic/foil bags or
               loose in a box.  Recall #F-739-4.
CODE           41223 and 41263.
MANUFACTURER   Horton's Downeast Foods, Inc., Waterboro, Maine.
RECALLED BY    Manufacturer, by telephone May 31, 1994, followed by letter
               June 1, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Maine, Florida, New York, Massachusetts.
QUANTITY       21.16 pounds were distributed.
REASON         Product is contaminated with Listeria monocytogenes.

_______________
PRODUCT        Cold Smoked Tuna, placed on foil/cardboard support vacuum
               packed in plastic, wholesale size 1-3 pounds and retail size
               4 ounces.  Recall #F-740-4.
CODE           Lot numbers:  11092, 11095, 911092, 121092, 911222, 910952.
MANUFACTURER   Horton's Downeast Foods, Waterboro, Maine.
RECALLED BY    Manufacturer, by telephone May 18 1994, followed by letter. 
               Firm-initiated recall complete.
DISTRIBUTION   Florida, Massachusetts, Maine, Missouri, North Carolina, New
               Hampshire, New York, Texas.
QUANTITY       196.55 pounds were distributed.
REASON         Product is contaminated with Listeria monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Keebler Zesty Meal Ground Saltine Crackers, in 15 pound
               cartons.  Recall #F-741-4.
CODE           Lot numbers:  080E, 081F, 081D, 081E, 082F, 082D, 083F,
               083E, 084F (manufactured 3/21-25/94).
MANUFACTURER   Keebler Company, Denver, Colorado.
RECALLED BY    Keebler Company, Elmhurst, Illinois, by letter June 21,
               1994, by fax and telephone June 22, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Iowa, Mississippi, New Jersey, Maryland, Pennsylvania,
               Florida, Indiana, Kentucky, Ohio, Tennessee, Texas,
               Minnesota, New Mexico.
QUANTITY       770 cases were distributed; firm estimates less than 20%
               remains on the market.
REASON         Product contains pieces of metal and plastic.

_______________
PRODUCT        Sutliff's Creme Candy Wafers, in 1 pound and 2 pound boxes
               in the following colors and flavors:  (a) Yellow; (b)
               Orchid; (c) Pink; (d) Apricot; (e) Ivory.  
               Recall #F-743/747-4.
CODE           None.
MANUFACTURER   Sutliff Candy Company, Inc., Seattle, Washington.
RECALLED BY    Manufacturer, by telephone June 15, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Washington state.
                                    -2-QUANTITY       Firm estimates none remains on the market.
REASON         Products contained undeclared FD&C Yellow #5.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Salted-In-Shell Sunflower Seeds, in 9.5 ounce plastic bags. 
               Recall #F-742-4.
CODE           13C4113, 14C4113, 17C4113, 18C4113, 19C4113, 21C4113,
               23C4113, 25C4113, 13P4113, 18P4113, 22P4113, 12R4113,
               15R4113, 18R4113, 21R4113, 23R4113, and 24R4113. 
MANUFACTURER   Sathers Inc., Round Lake, Minnesota (repacker).
RECALLED BY    Repacker, by telephone June 29, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and Chile.
QUANTITY       4,620 cases (12 bags per case) were distributed.
REASON         Product bears an incorrect ingredient statement.

_______________
PRODUCT        Sutliff Creme Candy Wafers, Blue, in 1 pound and 2 pound
               boxes.  Recall #F-748-4.
CODE           None.
MANUFACTURER   Sutliff Candy Company, Inc., Seattle, Washington.
RECALLED BY    Manufacturer, by telephone June 15, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Washington state.
QUANTITY       Firm estimates none remains on the market.
REASON         Product contains undeclared FD&C Blue #1.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Slo-Phyllin 80 mg Syrup (Theophylline), in 16 fluid ounce
               glass bottles, taken orally for asthmatic relief.  
               Recall #D-332-4.
CODE           Lot #89559 EXP 5/95.
MANUFACTURER   Rhone-Poulenc Rorer, Fort Washington, Pennsylvania.
RECALLED BY    Rhone-Poulenc Rorer, Collegeville, Pennsylvania, by letter
               June 13, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       7,931 bottles were distributed.
REASON         Potency is marginally above the upper specification limit of
               107%.

_______________
PRODUCT        Sterile Aerosol Talc Spray, in 4 g. cans, used for recurrent
               or persistent pneumothorax and recurrent or persistent
               pleural effusions.  Recall #D-333-4.
CODE           All lots.
MANUFACTURER   Talc de Luzenac, Cedex-France.
RECALLED BY    Bryan Corporation, Woburn, Massachusetts, by letter July 15,
               1994.  Firm-initiated recall ongoing.

                                    -3-DISTRIBUTION   Nationwide.
QUANTITY       Approximately 2,000 units.
REASON         Product marketed without new drug approval.

_______________
PRODUCT        Baker Norton brand Proglycem (diazoxide, USP) 50 mg
               capsules, in bottles of 100, used for treatment of chronic
               hypoglycemia.  Recall #D-334-4.
CODE           Lot #4C01871 EXP 3/96.
MANUFACTURER   Baker Norton Pharmaceuticals, Inc., Miami, Florida.
RECALLED BY    Manufacturer, by letter July 12, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       429 bottles were distributed.
REASON         Product does not meet content uniformity specifications.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Cardiotec Kit for the preparation of Technetium Tc 99m
               Teboroxime, a radioactive diagnostic.  Recall #D-330-4.
CODE           Lot 2F48395 of the formulation was packaged as three kits: 
               List #03206, lot #2F48395 (10 vials per kit);
               List #03205, lot #2F49441 (5 vials per kit);
               List #03207, lot #2F49452 (25 vials per kit).
MANUFACTURER   Bristol-Myers Squibb Company, New Brunswick, New Jersey.
RECALLED BY    Bristol-Myers Squibb Company, Princeton, New Jersey, by
               telephone December 9, 1993, followed by letter December 10,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       4,110 vials were distributed.
REASON         Vials may not meet the reconstituted physical appearance
               description as stated in the package insert.

_______________
PRODUCT        H.L. Moore brand Diazepam Tablets, USP 5 mg in bottles of
               100, indicated for the management of anxiety disorders or
               for the short-term relief of the symptoms of anxiety.  
               Recall #D-331-4.
CODE           Lot #3926-1014 EXP 4/96.
MANUFACTURER   Zenith Laboratories Caribe, Inc., Cidra, Puerto Rico.
RECALLED BY    Manufacturer, by telephone June 2, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Connecticut, Illinois.
QUANTITY       216 bottles were distributed; firm estimates none remains on
               the market.
REASON         A shipment erroneously contained package outserts for
               Prazosin.


                                    -4-
RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Tech Attach Connection Systems, EEG cables that can be
               purchased separately or as part of a Wyler Sphenoidal
               Electrode package.  Recall #Z-890-4.
CODE           All lots/codes of the Tech Attach Connection Systems,
               distributed prior to 12/1/93.  All models with unprotected
               electrode lead wires.
MANUFACTURER   Ad-Tech Medical Instrument Corporation, Racine, Wisconsin.
RECALLED BY    Manufacturer, by letter sent on or about December 1, 1993. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,368 cables.
REASON         Product labeling fails to provide adequate directions for
               use and the devices pose an unwarranted risk of injury to
               patients, because the products are not labeled to
               contraindicate use with apnea monitors.

_______________
PRODUCT        Siemens Conventional Mevatron Linear Accelerator System,
               equipped with an electron high dose rate (HI-E) mode and/or
               electron rotational therapy (EL-ROT) mode, used in the
               treatment of cancer with radiation.  Recall #Z-1098-4.
CODE           All lots.
MANUFACTURER   Siemens Medical Systems, Inc., Oncology Care Systems,
               Concord, California.
RECALLED BY    Manufacturer, by letter February 11, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 100 units.
REASON         In the Hi-E or EL-ROT mode, the display gives a decimal
               shift of one place to the right.  If not changed, the dose
               delivered will be ten times too much.

_______________
PRODUCT        MicroScan Rapid Positive Combo Type I, used for
               identification to species level and/or antimicrobial
               susceptibility of facultative and some fastidious aerobic
               gram-positive cocci and limonocytognes.  Recall #Z-1110-4.
CODE           Catalog #B1017-23, lot #09AUG93.
MANUFACTURER   Baxter Diagnostics, Microscan Division, West Sacramento,
               California.
RECALLED BY    Manufacturer, by telephone April 3, 1993, and letter April
               7, 1993. Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       347 boxes.
REASON         When Quality control testing was performed with E. faecalis
               ATCC 29212, Erythromycin demonstrated lower minimum
               inhibitory concentrations (MCI's) than the expected quality
               control range.

                                    -5-_______________
PRODUCT        Counter Recessed Three-Chamber Ultrasonic Cleaning System,
               Model E488, used to clean debris from dental instruments. 
               Recall #Z-1161-4.
CODE           All serial numbers beginning with 488.
MANUFACTURER   Esmadent Division, Esma Chemicals, Inc., South Holland,
               Illinois.
RECALLED BY    Manufacturer, by undated letter sent in October 1993.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       144 units were distributed.
REASON         Device appears to have component parts that are not properly
               grounded in the original factory configuration, which can
               result in electric shock and/or fire.

RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
=====================
_______________
PRODUCT        Microdase Disk - Modified Oxidase Test for Differentiation
               of Micrococci from Staphylococci.  Recall #Z-1111-4.
CODE           Lot #4711 EXP 5/31/94.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone December 15, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Nebraska, Missouri, Kansas, Oklahoma, New York, Georgia,
               Oregon, Kentucky, Puerto Rico, Canada, Japan.
QUANTITY       46 vials were distributed.
REASON         The device should have been made of ashless paper and was
               not.  Ashless paper carries less risk of interference with
               the test.

_______________
PRODUCT        Bacti-Disks, In Vitro Diagnostic, Microbiology Quality
               Control Organisms, which are stabilized, lyophilized,
               cultures of bacteria in solid disk form:  
               (a) Catalog #19-280; (b) Catalog #19-177; (c) Catalog # 19-
               052; (d) Catalog #19-060; (e) Catalog #19-035.  
               Recall #Z-1112/1116-4.
CODE           Lot numbers:  (a) 3330; (b) 2021; (c) 2340; (d) 3730; (e)
               4240.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone (a) January 18, 1993; (b) January
               31, 1994; (c) November 30, 1992; (d) November 3, 1992; (e)
               November 16, 1992.  Firm-initiated recall complete
DISTRIBUTION   (a) California, Pennsylvania, Ohio, New York, Florida, West
               Virginia; (b) Florida, New Jersey, New York, Michigan,
               Maryland, Wisconsin, Texas, Pennsylvania, South Carolina,
               Massachusetts; (c) Pennsylvania, California, New York,
               Florida; (d) Nationwide and Canada; (e) New Jersey,
               Massachusetts, North Carolina, Virginia, California,
               Australia.
QUANTITY       Firm estimates none remains on the market.

                                    -6-REASON         Some devices were contaminated and did not exhibit normal
               growth characteristics while other devices were mislabeled.

_______________
PRODUCT        C.A.T. Reagent Kit, in-vitro diagnostic reagent test for
               Chloramphenicol Acetyl-transferase.  Recall #Z-1117-4.
CODE           Lot #2141 EXP 2/28/94.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone December 29, 1993, and by fax
               December 30, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Australia, Canada.
QUANTITY       72 units were distributed between 4/23/93 and 12/29/93.
REASON         The control disk does not give the correct color reaction.

_______________
PRODUCT        P.R.A.S. Anaerobic Transport Medium:  (a) Anaerobic
               Transport Medium, used in the transportation of specimens
               containing anaerobic bacteria, Catalog #05-004 and 05-005;
               (b) Peptone Yeast (PY) Media with Lactose, used in the study
               of anaerobic bacteria, Catalog #05-105; (c) Peptone Yeast
               (PY) Media with Mannitol, used in the study of anaerobic
               bacteria, Catalog #05-109.  Recall #Z-1118/1120-4.
CODE           Lot numbers:  (a) All lots; (b) 2141; (c) 2231.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone (a) January 13, 1994; (b) January
               6, 1994; (c) January 4, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   (a) California, Virginia, Missouri, Georgia, Michigan,
               Colorado; (b) Massachusetts; (c) Pennsylvania.
QUANTITY       Firm estimates none remains on the market.
REASON         The proper color change did not occur when the media was
               exposed to air.

_______________
PRODUCT        (a) Thioglycollate Medium, w/o Indicator w/o Dextrose, 9 ml,
               used for the cultivation of aerobic and anaerobic bacteria,
               Catalog #06-4716; (b) Phenol Red Broth w/1% Melibiose, 5 ml,
               used in the determination of fermentation reactions of
               microorganisms, Catalog #06-2362; (c) PYR Broth 0.5 ml, used
               in the rapid, presumptive identification of Group A
               Streptococci, and Enterococci, Catalog #06-2085.  Recall #Z-
               1128/1130-4.
CODE           Lot numbers:  (a) 1031 EXP 12/31/93; (b) 4650 EXP 11/30/93;
               (c) 4230 EXP 10/31/93.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, (a) Product was discarded on or about May 18,
               1993; (b) by telephone.  Product was discarded on or about
               April 9, 1993; (c) by telephone January 12, 1993.  Product
               was discarded on or about January 2, 1993.  Firm-initiated
               recall complete. 
DISTRIBUTION   (a) Rhode Island; (b) Oklahoma, Washington D.C.; (c)
               Virginia, Colorado, Georgia, California, Nevada.

                                    -7-QUANTITY       (a) 400 tubes; (b) 60 tubes; (c) 340 tubes were distributed. 
               Firm estimates none remains on the market.
REASON         Thioglycollate Medium did not contain indicator, Phenol Red
               Broth was contaminated, and Pyr broth did not show a
               positive reaction.

_______________
PRODUCT        (a) Lap Disks, used for the detection of leucine
               aminopeptidase activity in streptococci and related genera,
               Catalog #21-129; (b) Bio-Ornithine Disk, used to biotype
               Haemophilus influenze and Haemophilus parinfluenzae, Catalog
               #21-024; (c) Pyr/Esculin Disk, used for rapid presumptive
               identification of Group A Streptococci and Enterococci,
               Catalog #21-138.  Recall #Z-1131/1133-4.
CODE           Lot numbers:  (a) 3531 EXP 8/31/94; (b) 3131 EXP 2/28/95;
               (c) 0521 EXP 8/31/94.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, (a) by telephone December 10, 1993.  Product
               was discarded on December 8, 1993; (b) by telephone December
               15, 1993.  Product was discarded on December 15, 1993; (c)
               by telephone May 12, 1993.  Product was discarded on or
               about May 4, 1993.  Firm-initiated recall complete.
DISTRIBUTION   (a) Arizona, New Jersey, Florida, Missouri, Kansas, North
               Carolina, Texas, California, Washington, D.C., Canada; (b)
               Wisconsin, Pennsylvania; (c) Nationwide, Australia, New
               Zealand.
QUANTITY       (a) 35 vials; (b) 3 vials; (c) 176 vials were distributed.
REASON         Known positive organisms failed to react or reacted in an
               atypical manner.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
==========
_______________
PRODUCT        Lewis Nursery Crumbles with CSP-250, medicated swine feed,
               50 pounds.  Recall #V-056-4.
CODE           Manufacturing dates:  2/1/94, 3/4/94, 3/30/94, 5/2/94,
               5/17/94.
MANUFACTURER   MFA Feed Mill, Kirksville, Missouri.
RECALLED BY    MFA Inc., Columbia, Missouri, by visit May 18, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Missouri.
QUANTITY       18 tons.
REASON         Product misformulated to contain only 50% of the drug
               ingredients.

_______________
PRODUCT        Orbenin-DC 7.5 G Syringe (Benzathine Cloxacillin) for
               intramammary infusion in cows.  Recall #V-057-4.
CODE           TC5833 EXP 3/95, TC5834 EXP 3/95, TC5847 EXP 3/95, TC5848
               EXP 3/95.
MANUFACTURER   Smithkline Beecham, Piscataway, New Jersey.
RECALLED BY    Manufacturer, by letter April 11, 1994.  Firm-initiated
               recall complete.
                                    -8-DISTRIBUTION   Nationwide.
QUANTITY       27,552 units of lot TC5833, 30,744 units of lot TC5834,
               32,004 units of lot TC5847, 28,152 units of lot TC5848 were
               distributed.
REASON         Four lots of the product are subpotent.


                                    -9-


END OF ENFORCEMENT REPORT FOR JULY 27, 1994.  BLANK PAGES MAY FOLLOW.
                                   ####