FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
07/30/1994

Recalls and Field Corrections:  Foods -- Class I -- 07/30/1994

July 20, 1994                                                 94-29
 
RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        Smoked Nova Chips, 5 pounds.  Recall #F-736-4.
CODE           None.
MANUFACTURER   Springfield Smoked Fish Company, Inc., Springfield,
               Massachusetts.
RECALLED BY    Manufacturer, by telephone June 1, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Connecticut.
QUANTITY       5 pounds were distributed.
REASON         Product was contaminated with Listeria monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT        Shark Cartilage Powder, bulk, 4.2 Kg.  Recall #F-388-4.
CODE           M-2368.
MANUFACTURER   Pharmline, Inc., Florida, New York.
RECALLED BY    Manufacturer, by letters of May 24 and 26, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   California, Canada, Japan.
QUANTITY       Firm estimates none remains on the market.
REASON         Product was contaminated with Salmonella.

_______________
PRODUCT        Various flavors of Estee Snack Crisps, waffle-shaped baked
               snacks packaged in both a small 0.66 ounce; 0.70 ounce for
               White Cheddar flavor only, and large 4.75 ounce flexible
               plastic bag:  (a) Chocolate; (b) Apple Cinnamon; (c) Lemon;
               (d) Ranch; (e) White Cheddar.  Recall #F-724/728-4.
CODE           All lots/date codes of products manufactured from 11/1/93
               through 3/15/94.
MANUFACTURER   The Estee Corporation, Parsippany, New Jersey.
RECALLED BY    Manufacturer, by letters dated March 17 and 21, 1994, and
               May 4, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Mexico.
QUANTITY       (a) 4,454 cases; (b) 4,502 cases; (c) 5,005 cases; (d) 3,509
               cases; (e) 238 cases were distributed.  As of 4/22/94, firm
               estimated that less than 3% of product remained on the
               market.
REASON         Products may contain metal fragments.

_______________
PRODUCT        Dark Chocolate and Milk Chocolate Non Pareils in 8 ounce
               boxes.  Recall #F-729/730-4.
CODE           None.
MANUFACTURER   Criterion Chocolates, Inc., Eatontown, New Jersey.
RECALLED BY    Manufacturer, by letter January 13, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   New Jersey.
QUANTITY       816 boxes were distributed.
REASON         Products contain undeclared FD&C Yellow No. 5.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT        Various butter products:
               (a) Level Valley 47-cut Lightly Salted Butter, single
               serving size;
               (b) Monarch 47-cut Lightly Salted Butter, single serving
               size;
               (c) Level Valley Fresh Sweet Cream Butter, 4 quarters, 16
               ounces;
               (d) Kroger Lightly Salted Butter, 4 quarters, 16 ounces;
               (e) Shur Fresh Lightly Salted Sweet Cream Butter, 4
               quarters, 16 ounces.  Recall #F-731/735-4.
CODE           (a) Recalled codes begin with 482; (b) Recalled codes begin
               with 483; (c) Pull date 5/4/94, churn number embossed on the
               end of the package is 483; (d) Pull date 5/7/94, churn
               number 483; (e) Pull date 5/7/94, churn number 483.
MANUFACTURER   Level Valley Dairy Company, West Bend, Wisconsin.
RECALLED BY    Manufacturer, by letter March 7, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Tennessee, Florida, Georgia, North Carolina, South Carolina.
QUANTITY       (a) 558 cases (17 pounds per case); (b) 558 cases (17 pounds
               per case); (c-e) 228 cases (36 pounds per case) of the 16
               ounce packages of butter quarters.
                                    -2-REASON         Products contain dieldrin at a level greater than the
               established tolerance.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Lindane Lotion, USP 1%, Rx topical, used for the treatment
               scabies, in 1 pint bottles sold under the following labels: 
               Pennex, PBI, Goldline, GG, Major, Qualitest, Rugby, Schein. 
               Recall #D-327-4.
CODE           Lot numbers:  19490A, 19490E, 19516A, 19806A, 19806C,
               19806E, 19806F, 19866A, 19866C, 19878P, 19878R, 19878T,
               19878X.
MANUFACTURER   Pennex Pharmaceuticals, Inc., formerly known as
               Pharmaceutical Basics, Inc., Morton Grove, Illinois.
RECALLED BY    Manufacturer, by letter dated June 29, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       28,856 bottles were distributed; firm estimates 25% of the
               product remains on the market.
REASON         Lots 19490, 19516 and 19806 do not meet content uniformity
               specification.  Lots 19866 and 19878 are superpotent.

_______________
PRODUCT        Salsalate Tablets, USP:  (a) 500 mg, in 100 and 1000 count
               bottles; (b) 750 mg, in 100 and 1000 count bottles.  
               Recall #D-328/329-4.
CODE           Lot numbers:  (a) 3FE01A, 3FE01B, 3AP02A, 3AP02B; (b)
               3FE02A, 3FE02B, 3AP01A, 3AP01B.
MANUFACTURER   Caraco Pharmaceutical Laboratories, Inc., Detroit, Michigan.
RECALLED BY    Manufacturer, by letter July 20, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Colorado, Missouri, Arkansas, Illinois, Michigan, Tennessee,
               Louisiana, Florida.
QUANTITY       (a) 85 bottles of 100 of lots 3FE01B and 3AP02B; 26 bottles
               of 1,000 of lot 3FE01A; (b) 62 bottles of 100 of lot 3FE02B;
               and 316 bottles of 1,000 of lots 3AP01A and 3FE02A.
REASON         Products do not meet dissolution specifications.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Blood Bank Information System (BBIS) Computer Software. 
               Recall #B-316-4.
CODE           All versions released prior to version 4.
MANUFACTURER   Collaborative Medical Systems, Inc., Waltham, Massachusetts.
RECALLED BY    Manufacturer, by letters of June 25 and 30, 1993.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Iowa, Massachusetts, Oklahoma.
QUANTITY       5 systems.

                                    -3-REASON         Software, that contained programming errors which could
               potentially result in the release of unsuitable blood
               products, was distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        (a) EF Synergy Quad Plated Media, used as a predictor of
               synergy of an aminoglycoside and penicillin when testing
               enterococci isolates;
               (b) Chocolate Agar Mono Plated Media, used in the culturing
               of Pathogenic Neisseria and Haemophilus.  
               Recall #Z-954/955-4.
CODE           Lot numbers:  (a) 2721 EXP 8/31/93, Catalog #02-454; (b)
               3540 EXP 11/12/92, Catalog #01-302.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone September 29, 1992.  Firm-
               initiated recall complete.
DISTRIBUTION   (a) Arkansas, Maine, Massachusetts, New Hampshire, New York,
               Rhode Island, California, Ohio, Michigan, Utah, Nebraska,
               Missouri, Florida, Virginia, Pennsylvania; (b) California,
               Utah, New Mexico.
QUANTITY       (a) 660 plates; (b) 1,700 plates were distributed.  Firm
               estimates none remains on the market.
REASON         The control quadrant for the EF Synergy Quad Plated Media
               failed to support the growth of the challenge organisms used
               as positive controls; and the quality control organism,
               Haemophilus influenzae, for the Chocolate Agar Mono Plated
               Media did not grow on this lot of product.

_______________
PRODUCT        Pro-Lok Beech brand, Shielded Needle Connector, Needle
               Housing with Removable Needle, 18GA x 5/8 (16 mm), sterile:
               (a) Catalog #01-9018; (b) Catalog #01-9020.  
               Recall #Z-983/984-4.
CODE           All lots with prefix AE through CC, which represents all
               lots manufactured between 5/91 and 3/93.
MANUFACTURER   VLV Associates, Inc., East Hanover, New Jersey.
RECALLED BY    Beech Medical Products, Newtown, Pennsylvania, by letter
               August 28, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania, Maryland, New York, Massachusetts, Illinois,
               Texas, Michigan, California, Ohio, Colorado, Missouri,
               Washington state, Minnesota, Indiana.
QUANTITY       Approximately 2 million plus units were distributed; firm
               estimates none remains on the market.
REASON         The female luer of the Pro-Lok housing cracks and may leak
               when used with certain IV solutions.

_______________
PRODUCT        CardioPump ACD (Active Compressed/Decompression)
               Resuscitator both prototype and production models.  
               Recall #Z-1078-4.

                                    -4-CODE           All codes including lot numbers 262237 and 262242.
MANUFACTURER   Ambu International A/S, Glostrup, Denmark.
RECALLED BY    Ambu, Inc., Linthicum, Maryland, by telephone and letter
               April 19, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Minnesota, New York.
QUANTITY       506 devices.
REASON         The firm was using data from an unapproved clinical
               investigation on human subjects to support a 510(k)
               application.

_______________
PRODUCT        Storz Intraocular Lens, Model P010UV.  Recall #Z-1080-4.
CODE           Serial numbers:  1G1705, 1G1710, 1G1717 through 1G1719,
               1G1721 through 1G1727, 1G1730, 1G1731, 1G1733, 1G1734,
               1G1737, 1G1740, 1G1745, 1G1747, 1G3811.
MANUFACTURER   Storz Ophthalmics, Inc., Clearwater, Florida.
RECALLED BY    Manufacturer, by letter April 20, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Ohio, Florida, California, Singapore, Germany, Brazil,
               Guatemala.
QUANTITY       21 lenses remain in commerce that may still be unimplanted.
REASON         The lenses were mislabeled  with the incorrect refractive
               index.

_______________
PRODUCT        Microvasive Ablaser Side-Firing Laser Guide, a laser
               delivery guiding catheter, indicated for use for soft tissue
               ablation and hemostasis treatment in urological conditions. 
               Recall #Z-1088-4.
CODE           Catalog #800-102, lot numbers:  EM02061 EXP 12/95, 118340
               EXP 2/96, 118165 EXP 2/96, 118566 EXP 2/96, 119078 EXP 2/96,
               119062 EXP 2/96.
MANUFACTURER   Boston Scientific Corporation, Milford, Massachusetts.
RECALLED BY    Boston Scientific Corporation, Watertown, Massachusetts,
               March 22, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Michigan, California, Missouri, Mississippi, Illinois, New
               York, Kentucky, Virginia, Tennessee, Alabama, Indiana,
               Massachusetts, Pennsylvania, Colorado, Florida, Georgia,
               Washington, D.C.
QUANTITY       179 units were distributed; firm estimates none remains on
               the market.
REASON         The gold plated catheter tip may detach during surgery
               causing the tip to remain within the body or the laser
               reflecting mirror in the tip may char and burn tissue.

_______________
PRODUCT        Disposable reservoir bags for Infumed and Walkmed infusion
               pumps: 
               (a) IPR-86 Reservoir Bag, 65 ml;
               (b) IPR-150 Reservoir Bag, 150 ml; 
               (c) IPR-250 Reservoir Bag, 250 ml; 
               (d) MF-65 Reservoir Bag, 65 ml;   
               (e) PR-70 Reservoir Bag, 65 ml Recall #Z-1089/1093-4.
                                    -5-CODE:          Lot numbers:  (a) MF91435, MF91483, MF91487, MF91505,
               MF911516, MF92011, MF92069, MF92077, MF92088, MF92114,
               MF92138, MF92186, MF92441E, MF92273E, MF92274E, MF922851 and
               MF92328; 
               (b) MF91421, MF91513, MF92013, MF92033, MF92065, MF92080,
               MF92117, MF92125, MF92159, MF92204E, MF92243E and MF922761; 
               (c) MF91454, MF91510, MF92015, MF92067, MF92210E, MF92244E, 
               MF922791 and MF92330; 
               (d) MF91445, MF92017  and MF92091; 
               (e) MF92024. 
MANUFACTURER   Medex, Inc., Duluth, Georgia.
RECALLED BY    Manufacturer, by letter June 11, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, United Kingdom, Germany.
QUANTITY       171,010 bags.
REASON         The female luer lock will crack when being used preventing
               administration of proper amount of prescribed fluids.

_______________
PRODUCT        Baxter/Travenol Fenwal Blood Warmer, used to warm blood
               products and intravenous fluids before administration to
               reduce complications associated with the infusion of cold
               solutions:  (a) Product #4R4305 - 115 volts (U.S.
               distribution); (b) Product #4R4306 - 220 volts
               (international distribution).  Recall #Z-1094/1095-4.
CODE           Serial numbers 15821 through 19320, and those units which
               have had their housing replaced with blue housings with
               black EMI coating on the inside.
MANUFACTURER   Panlmatic Company, Elk Grove Village, Illinois.
RECALLED BY    Baxter Healthcare Corporation, Deerfield, Illinois, by
               letter June 15, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Colombia, Brazil, Venezuela, United
               Kingdom.
QUANTITY       An estimated 3,500 units are affected and remain on the
               market.  These affected units were distributed between May
               1985 and September 1987.
REASON         There is a potential for electrical shock if the wiring were
               to come loose and contact the electrical magnetic
               interference (EMI) coating inside the housing.

_______________
PRODUCT        CCS Maestro II Models 231, 232, 325, and 333 Cardiac
               Pacemakers.  Recall #Z-1099/1102-4.
CODE           All serial numbers.
MANUFACTURER   Cardiac Control Systems, Inc., Palm Coast, Florida.
RECALLED BY    Manufacturer, by letter March 25, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Arkansas, California, Connecticut, Florida, Louisiana,
               Massachusetts, Maine, Montana, North Carolina, New Jersey,
               New York, Ohio, Pennsylvania, Rhode Island, South Carolina,
               Tennessee, Texas, Wisconsin. 

                                    -6-QUANTITY       416 pacers were distributed domestically.  Of those, 284
               were implanted and 132 were recovered (per final report).  A
               total of 574 pacers were distributed internationally and all
               were implanted.
REASON         An intended future safety feature included in the hybrid
               circuitry of the present device design can be spontaneously
               activated in some units, causing the device to act as an
               asynchronous single-chamber pacer.  The likely result of
               such an event would be that the pacer would revert to an 80
               pulses per minute (ppm) mode, which may not be a
               satisfactory default if the mode persists.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Beta Lysin Reagent, 5 ml, a liquid microbial reagent for the
               identification of Strep. agalactiae.  Recall #Z-956-4.
CODE           Catalog #21-022, lot #4111 EXP 10/31/94.
MANUFACTURER   Remel Limited partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone January 20, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Colorado, Illinois, Oklahoma.
QUANTITY       15 units were distributed from 10/11/93 to 1/20/94.
REASON         The expected positive QC reaction, an enhanced crescent of
               beta hemolysis with Strep. Agalactiae, was not occurring.

_______________
PRODUCT        McFarland Equivalence Turbidity Standards 4.0, used as
               standards in adjusting densities of bacterial suspensions. 
               Recall #Z-957-4.
CODE           Catalog #20-414, Lot #0154 EXP 4/4/94.
MANUFACTURER   Scientific Device Lab., Glenview, Illinois.
RECALLED BY    Remel Limited Partnership, Lenexa, Kansas, by telephone
               October 20, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Oklahoma, Texas, Florida, New Jersey, Washington state,
               Michigan, Kansas, Missouri, Illinois, Colorado, New Mexico,
               New York, Nebraska, Guam, Canada.
QUANTITY       37 units were distributed.
REASON         The individual tubes were incorrectly labeled.

_______________
PRODUCT        Piperacilin 100 mcg Susceptibility Test Disks, for in-vitro
               diagnostic use.  Recall #Z-958-4.
CODE           Catalog #30-813, Lot #2830 EXP 7/31/94.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone and letter May 7, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 386 units were distributed; firm estimates
               none remains on the market.
REASON         The individual cartridge labeled Piperacillin was
               incorrectly labeled.

                                    -7-_______________
PRODUCT        Virocult ENT, in-vitro diagnostic collection and transport
               system.  Recall #Z-959-4.
CODE           Catalog #61-9755, Lot #3811 EXP 2/28/95.
MANUFACTURER   Medical Wiring & Equipment, Sparta, New Jersey.
RECALLED BY    Remel Limited Partnership, Lenexa, Kansas, by letter January
               20, 1994.  Firm-initiated recall complete.
DISTRIBUTION   New York, Ohio, Kansas.
QUANTITY       8 units of 125 swabs were distributed.
REASON         The product had an incorrect expiration date printed on the
               label of the outer container.

_______________
PRODUCT        Remel RPR Liquid Controls, to be used with the Remel RPR
               Card Test Kit.  Recall #Z-960-4.
CODE           Catalog #16-307, Lot #3083 EXP 4/30/94.
MANUFACTURER   Ampcor, Camden, New Jersey.
RECALLED BY    Remel Limited Partnership, Lenexa, Kansas, by telephone May
               27, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Illinois, Missouri, Iowa, Tennessee, Michigan, Alabama.
QUANTITY       20 units were distributed.
REASON         The reactive control contained black specs and turbidity.

_______________
PRODUCT        Contact I Sabouraud Dextrose Agar with Polysorbate 80 with
               Lecithin, used in the detection and enumeration of
               microorganisms present on surfaces of sanitary importance. 
               Recall #Z-961-4.
CODE           Catalog #11-1306, Lot #3641 EXP 12/9/93.
MANUFACTURER   Remel Limited Partnership, Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone October 19 0r 20, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   California.
QUANTITY       500 plates were distributed; firm estimates none remains on
               the market.
REASON         The product was incorrectly labeled as TSA instead of SAB-
               DEX.

_______________
PRODUCT        Linvatec Cannuloc, threaded cannula, used during orthoscopic
               surgery to introduce various surgical instruments into the
               surgical site.  Recall #Z-1079-4.
CODE           Catalog #9822, Lot #47855.
MANUFACTURER   Linvatec, Division of Bristol-Myers Squibb, Largo, Florida.
RECALLED BY    Manufacturer, by letter April 6, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Georgia, Massachusetts, Minnesota, Maine, Denmark, Canada.
QUANTITY       24 boxes of 10 were distributed.
REASON         75 mm Cannuloc Threaded Cannulas were labeled as 50 mm
               Cannuloc Threaded Cannulas.

                                    -8-_______________
PRODUCT        Wall Mount Brackets, ABS construction, part #008-0046-00, an
               accessory item for PROPAQ Patient Monitor, Models 102, 104,
               and 106.  Recall #Z-1083-4.
CODE           All brackets with part #008-0046-00 ABS plastic
               construction.
MANUFACTURER   Protocol Systems, Inc., Beaverton, Oregon.
RECALLED BY    Manufacturer, by letters in January/February 1993, and
               May/June 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       196 units were distributed.
REASON         An environment of high vibration or severe force could cause
               fracture of the plastic portion of the bracket.

_______________
UPDATE         2FR V-Cath Introducer Needle, Recall #Z-683-4, manufactured
               by HDC Corporation, which appeared in the April 27, 1994,
               Enforcement Report should be identified as Product #350-20,
               Lot #1013.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
===========
_______________
PRODUCT        (a) Organic Iodide with Dextrose, in 1 pound and 25 pound
               containers; (b) Organic Iodide with Saline, in 1 pound and
               25 pound containers.  Recall #V-054/055-4.
CODE           Lot numbers:  (a) WAB50575, WAB50580; (b) WAB50590,
               WAB50595.
MANUFACTURER   Chem-Tronics, Inc., Leavenworth, Kansas.
RECALLED BY    The Butler Company, Dublin, Ohio, by letter April 21 and 25,
               1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       6,152 units were distributed.
REASON         Labeled directions for use in cattle for ethylenediamine
               dihydroiodide (EDDI) exceeds the daily recommended dietary
               requirements for cattle.

                                 -9-

END OF ENFORCEMENT REPORT FOR JULY 20, 1994.  BLANK PAGES MAY FOLLOW.

                                   ####