FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
05/18/1994

Recalls and Field Corrections:  Drugs -- Class II -- 05/18/1994

May 18, 1994                                                    94-20

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Transfilled Compressed Medical Oxygen in D & E size
               cylinders.  Recall #D-274-4.
CODE           All fill dates.
MANUFACTURER   ADR Home Care Services, East Setauket, New York.
RECALLED BY    Manufacturer, by telephone and letter April 25, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   New York.
QUANTITY       53 E and 17 D cylinders were distributed.
REASON         Current good manufacturing practice deficiencies.

_______________
PRODUCT        (a) Loestrin FE 1/20 Tablets, Norethindrone Acetate and
               Ethinyl Estradiol Tablets, USP and Ferrous Fumarate Tablets,
               USP, (5 packages of 28 tablets); 
               (b) Loestrin 1/20 Tablets, Norethindrone Acetate and Ethinyl
               Estradiol Tablets, USP, (5 packages of 21 tablets).
               Recall #D-275/276-4.
CODE           Lot numbers:  (a) 05632FC EXP 3/95, 05632FG EXP 2/95;
               (b) 05632FA EXP 3/95, 05632FB EXP 3/95, 05632FF EXP 3/95.
MANUFACTURER   Warner Lambert, Fajardo, Puerto Rico.
RECALLED BY    Parke-Davis, Division of Warner Lambert Company, Morris
               Plains, New Jersey, by letter January 17, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       Approximately 63,284 units were distributed in the United
               States and 56,540 units in Canada.
REASON         Products did not meet content uniformity specifications.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        Recovered Plasma.  Recall #B-268-4.
CODE           Unit #93-2300.
MANUFACTURER   St. Joseph Hospital Blood Bank, Reading, Pennsylvania.
RECALLED BY    Manufacturer, by telephone December 3, 1993, followed by
               letter December 13, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit.
REASON         Blood product, untested for viral markers, was distributed.

_______________
PRODUCT        (a) Whole Blood; (b) Platelets.  Recall #B-269/270-4.
CODE           Unit numbers:  (a) 2067539; (b) 1743562.
MANUFACTURER   LifeSource, Glenview, Illinois.
RECALLED BY    Manufacturer, by telephone and by Fax February 25, 1994,
               followed by letter March 9, 1994.  Firm-initiated recall
               complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit of each component.
REASON         Blood products, which tested repeatedly reactive for the
               antibody to hepatitis B core antigen (anti-HBc), were
               distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Advantx CFM Fluoroscopic X-Ray Imaging System Positioning
               Arm, Model numbers 45200707, 45554027, 45554448, 45553680,
               and 45561622.  Recall #Z-373/377-4.
CODE           Model number          serial number 
               45200707                3059000 
               45200707                3058400 
               45200707                7156401 
               45200707                7458100 
               45554027                02157BU6 
               45554027                17134BU8 
               45554027                11830BU7 
               45554027                15373BU4 
               45554027                08727BU0 
               45554027                16813BU8 
               45554027                04870BU2 
               45554027                10208BU7 
               45554027                11961BU0 
               45554027                13534BU3 
               45554027                7478601 

                                    -2-               45554448                9963801 
               45554448                99305 
               45553680                45553680 
               45561622                08837BU7 
MANUFACTURER   GE Medical Systems - Europe, Buc (Paris), France.
RECALLED BY    General Electric Company, Medical Systems Division,
               Waukesha, Wisconsin, by Urgent Safety Notice sent February
               17, 1994.  Firm-initiated recall complete.
DISTRIBUTION   California, Florida, Indiana, Kentucky, Massachusetts,
               Michigan, Ohio, Pennsylvania, Tennessee, Texas, Argentina,
               Chile.  
QUANTITY       19 units.  
REASON         An improper weldment on the arm structure for the image
               intensifier may crack and fail permitting the image
               intensifier to separate from its support and fall.

_______________
PRODUCT        Manometric Catheters:
               (a) Esophageal Catheters Catalog No. EMC-3; 
               (b) Esophageal Catheters Catalog No. EMC-4; 
               (c) Esophageal Catheters Catalog No. EMC-6; 
               (d) Esophageal Catheters Catalog No. EMC-8; 
               (e) Anorectal Catheters Catalog No. AMC-4; 
               (f) Anorectal Catheters Catalog No. AMC-8; 
               (g) Vector Volume Catheters Catalog No. VMC-4; 
               (h) Vector Volume Catheters Catalog No. VMC-6; 
               (i) Vector Volume Catheters Catalog No. VMC-8. 
               Recall #Z-482/490-4.
CODE           Serial numbers:  ZX00-ZR99.  All catheters throughout these
               serial codes except ZW44 and ZU62 which were removed from
               stock and scrapped.
MANUFACTURER   Zinetics, Salt Lake City, Utah.
RECALLED BY    Manufacturer, by letter October 26, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas, Colorado, Idaho, Pennsylvania, Illinois, South
               Carolina, New Mexico, Washington state, California,
               Michigan, New York, Indiana, Utah, Wyoming, Missouri,
               Maryland.
QUANTITY       700 units.
REASON         The depth reference markings can be partially or completely
               removed during reuse.  

_______________
PRODUCT        Sterile Nealon Surgical Gloves with 100% Lactose Powder:
               (a) Catalog No. 376655, Size 5-1/2 
               (b) Catalog No. 376660, Size 6
               (c) Catalog No. 376665, Size 6-1/2 
               (d) Catalog No. 376670, Size 7
               (e) Catalog No. 376675, Size 7-1/2
               (f) Catalog No. 376680, Size 8
               (g) Catalog No. 376685, Size 8-1/2
               (h) Catalog No. 376690, Size 9. Recall #Z-512/519-4.

                                    -3-CODE           Last four digits of their lot number is less than 3279.
MANUFACTURER   Becton Dickinson AcuteCare, Honed Path, South Carolina;
               B-D Acute Care Plant, El Paso, Texas;
               Elamex Company, Juarez, Mexico (related subsidiary).
RECALLED BY    Becton Dickinson & Co., Franklin Lakes, New Jersey by letter
               November 4, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       714,800 pairs were distributed; firm estimates 150,000 pairs
               remain on the market.
REASON         Products are being manufactured and distributed without an
               approved PMA for the 100% lactose powder donning agent.

______________
PRODUCT        All unprotected patient cables and lead wires provided with
               Corometrics Medical Systems brand Apnea Monitors. The
               following products are affected:
               Catalog No:      Description:
               1305AAO          Model 502 Patient Cable
               1306AAO          Model 502 Lead Wires
               1307AAO          Model 502 (IEC) Lead Wires
               1308AAO          Model 556 Patient Cable
               1309AAO/CAO      Model 556 Lead Wires
               1408AAO          Model 505/506/515 Patient Cable
               1408BAO          Model 505/506/515 Patient Cable
               2906AAO/BAO      Electrodes w/Preattached Lead Wires
               2906CAO          Preemie Electrodes w/Preattached Lead Wires
               5109AAO          Model 505/506/515 Lead Wires
               5109BAO          Model 505/5-6/515 Lead Wires.
               Recall #Z-544-4.
CODE           All models with unprotected (exposed pin style) lead wires
               and unprotected patient cables.
MANUFACTURER   Corometrics Medical Systems, Inc., Wallingford, Connecticut.
RECALLED BY    Manufacturer, by letter dated October 11, 1993. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Undetermined.
REASON         Device labeling fails to provide adequate directions for use
               and the devices pose an unwarranted risk of injury to
               patients, because the product is not labeled to warn against
               the hazards of unprotected electrode lead wires being
               connected to electrical power sources.

_______________
PRODUCT        G4A electrode lead wires used with model EGS 300 Portable
               High Voltage Galvanic Stimulator.  Recall #Z-547-4.
CODE           All codes distributed from January 1991 to January 1994.
MANUFACTURER   Electro-Med Health Industries, North Miami, Florida.
RECALLED BY    Manufacturer, by letter dated February 10, 1994.
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 1500 cable sets.

                                    -4-REASON         Product labeling fails to provide adequate directions for
               use and the devices pose an unwarranted risk of injury to
               patients, because the products are not labeled to
               contraindicate use with apnea monitors.

_______________
PRODUCT        Cobe Spectra Apheresis System - Pump Rotor, used for the
               collection of blood products:
               (a) Catalog No. 950000-000
               (b) Catalog No. 950000-001
               (c) Catalog No. 950000-002
               (d) Catalog No. 950000-003
               (e) Catalog No. 950000-004
               (f) Catalog No. 950000-005
               (g) Catalog No. 950000-006
               (h) Catalog No. 950000-007
               (i) Catalog No. 950000-008.  Recall #Z-711/719-4. 
MANUFACTURER   Cobe BCT, Incorporated, Lakewood, Colorado.
RECALLED BY    Manufacturer, by letter January 19, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       227 instruments.
REASON         The adjustment screw which provides for proper tubing
               occlusion and rotor tension was not assembled correctly and
               may cause delivery of the anticoagulant at levels greater
               than expected.

_______________
PRODUCT        Hardware Trigger Board, part #0005-1030, for console in BVS
               5000 Bi-ventricular Support System.  Recall #Z-732-4.
CODE           Console serial numbers 1121-1150, 1154, 1157-1161 and two
               hardware trigger boards serial numbers 0992-E0009 and 0992-
               E0010 which were not in consoles and put into spare parts
               kits.
MANUFACTURER   Electronic Associates, Inc. (EAI), West Long Branch, New
               Jersey (Hardware Trigger board part #0005-1030);
               Abiomed, Inc., Danvers, Massachusetts (Bi-Ventricular
               Support Console part #0005-9007).
RECALLED BY    Abiomed, Inc., Danvers, Massachusetts, by letter June 16,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Illinois, Ohio, Missouri, New York, North
               Carolina, Pennsylvania, Wisconsin, Austria.
QUANTITY       15 boards in console and 2 in spare parts kits.
REASON         Capacitors in the hardware trigger board may short causing
               the board to fail and the console to shut down.

_______________
PRODUCT        Baxter Pharmaseal Aquamatic K-Modules, Model K-20 Series; a
               small reservoir/heater/pump used to warm and circulate water
               for a heat therapy pad:
               (a) Catalog #S64N00100
               (b) Catalog #S64N02000

                                    -5-               (c) Catalog #S11185-FRE
               (d) Catalog #S64N02500
               (e) Catalog #S64N03000
               (f) Catalog #S64N03500
               (g) Catalog #S64N04000.  Recall #Z-733/739-4.
CODE           All serial numbers starting with 10001S through 14470S
               serial numbers: T00671 through T01170, T01321 through
               T01820, T03091 through T03390, T04891 through T05190, T06191
               through T06490, T06991 through T07590, T08205 through
               T08534, T09035 through T09334, T10335 through T11144, T12145
               through T12644, T13680 through T16759.
MANUFACTURER   (a) Baxter Healthcare, Savage, Maryland (plant is now
               closed). 
               (b) Baxter Healthcare, Largo, Florida. 
               (c) Bront Machinery, Dayton, Ohio. (sub-assembly
               manufacturer, most responsible for problem).
RECALLED BY    Baxter Healthcare Corporation, Convertors/Custom
               Sterile/Pharmaseal Division, Deerfield, Illinois, by letter
               November 15, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       11,903 units were distributed.
REASON         After repeated use, the solder affixing the heating element
               to the reservoir could fail, and could result in the
               potential for the heating element to come into contact with
               the device's plastic housing, potentially causing smoke
               and/or melting of the plastic housing.

_______________
PRODUCT        Inflation Pro II Model IP 9001, used to inflate
               cardiovascular balloon catheters for PCTA procedures. 
               Recall #Z-745-4.
CODE           All lots with the first two digits of the of number reading
               "2A" through "2L" and "3A" through "3L".
MANUFACTURER   Edwards Critical-Care, Interventional Cardiology Division,
               Irvine, California.
RECALLED BY    Manufacturer, by letter January 24, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       86,002 units.
REASON         The balloon may oversize or rupture due to a defective
               gauge.

_______________
PRODUCT        Disposable Vinyl Examination Gloves, 100 pieces per carton. 
               Recall #Z-747-4.
CODE           LCL Taiwan & Company, Taipei, Taiwan.
RECALLED BY    C&K Manufacturing and Sales Company, Westlake, Ohio, by
               letter June 28, 1993, followed by telephone June 29, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,469 cases were distributed.

                                    -6-REASON         The gloves contain more holes than allowed by FDA's
               Compliance Policy Guide 7424.31.

_______________
PRODUCT        Tooth Conditioner Gel, in 3 ml plastic syringes.  
               Recall #Z-749-4.
CODE           All product manufactured since 1986 to October 1993.
MANUFACTURER   L.D. Caulk, Division of Dentsply International, Milford,
               Delaware.
RECALLED BY    Manufacturer, by letter May 3, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       922 units.
REASON         The syringe may split or explode when the user attempts to
               eject the contents (34% phosphoric acid gel) or the tip may
               clog which could cause the product to shoot out of the tip
               of the syringe in an uncontrolled manner when the clog is
               overcome.

_______________
PRODUCT        Thora-Seal I, Plastic Underwater Chest Drainage System. 
               Recall #Z-750-4.
CODE           Lot #632104 or lower.
MANUFACTURER   Sherwood Medical Company, St. Louis, Missouri.
RECALLED BY    Manufacturer, by letter October 13, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       19,500 units.
REASON         When water is added to the fill line marked on the bottle,
               the opening at the bottom of the water seal tube may not be
               completely immersed in water as required, which may result
               in air entering the chest cavity.

_______________
PRODUCT        Medfusion Model 2001 and 2010 Syringe Infusion Pumps. 
               Recall #Z-751/752-4.
CODE           Serial numbers 9900 to 13400.
MANUFACTURER   Medfusion, Inc., Duluth Georgia.
RECALLED BY    Manufacturer, by letter on or about November 2, 1992.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide, Canada, England, Germany.
QUANTITY       3,500 units were distributed.
REASON         Use of incorrect inserts (screws) placed in the bottom
               housing of the infusion pump which may become loose
               completely detached, causing the pump to fail.

_______________
PRODUCT        Sci-O-Tech Full Electric and Semi-Electric Adjustable
               Hospital Beds, Catalog 1000 series and semi-electric
               adjustable hospital beds and the electric control boxes for
               the hand controls on the beds, which raise and lower head
               and foot sections, and the entire bed:

                                    -7-               (a) Semi-Foot Section, Part #9541125
               (b) Electric Foot Section, Part #9541120
               (c) 2-Motor Control Box, Part #5685552
               (d) 3-Motor Control Box, Part #5685554.
               Recall #Z-759/762-4.  
CODE           (a) Part #9541125; (b) Part #9541120; (c) Part #5685552;
               (d) Part #5685554.
MANUFACTURER   Sci-O-Tech, Inc., Lancaster, Pennsylvania.
RECALLED BY    Manufacturer, by letter.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 15,000 beds were distributed.
REASON         Beds foot and head sections may raise or lower suddenly
               without warning.

_______________
PRODUCT        Cobe Optima HVRF (Hollow Fiber Oxygenators), Catalog
               #050212-000.  Recall #Z-763-4.
CODE           02Z7000.
MANUFACTURER   Cobe Cardiovascular, Inc., Arvada, Colorado.
RECALLED BY    Manufacturer, by telephone beginning February 25, 1994,
               followed by letter March 1, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Michigan, New York, New Jersey, North Dakota, Oklahoma,
               Ohio.
QUANTITY       112 units.
REASON         A defect exists in the heat exchangers which may make them
               prone to premature failure under severe pulsatile flow
               conditions, resulting in small water to blood leaks.  

_______________
PRODUCT        Shampaine Surgical Tables:  (a) Model 4900B, Battery Powered
               Surgical Tables; (b) Model 4900E, Mains Operated Surgical
               Tables.  Recall #Z-790/791-4.
CODE           All manufactured between October 1992 through December 1993.
MANUFACTURER   MDT Diagnostic Company, North Charleston, South Carolina.
RECALLED BY    Manufacturer, by letter February 4, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Greece, Hong Kong, Korea.
QUANTITY       289 units.
REASON         A potential manufacturing defect by a component supplier may
               result in the tables failing to function as intended.

_______________
UPDATE         HAID Universal Bone Plate (UBP) System, Recall #Z-755/758-4,
               which appeared in the May 11, 1994 Enforcement Report should
               read:
               (d) Part #650005-005, Catalog No. S/L 50-4551 (Evaluation
               Set).

                                    -8-
RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Raichem Calcium Reagent Standard Reagent found in Raichem
               Calcium Reagent In-vitro Diagnostic Kits:
               (a) Catalog #84104; (b) Catalog #84105; (c) Catalog #85188;
               (d) Catalog #85210.  Recall #Z-726/729-4.
CODE           All kits with lot numbers 54106, 55010, 06011, 58022.
MANUFACTURER   Reagents Applications, Inc., San Diego, California.
RECALLED BY    Manufacturer, by telephone June 15, 1993.  Firm-initiated
               recall ongoing.
DISTRIBUTION   California, Idaho, Illinois, Massachusetts, Nebraska, New
               York, Wisconsin, Puerto Rico, Israel, Mexico.
QUANTITY       119 vials were distributed.
REASON         Defective closures (vial stoppers) are uncoated on the
               standard and may lead to leaching of contents causing
               inaccurate test results.

_______________
PRODUCT        Versacath Introduction Trays, for introduction of Versacath
               Mainline Central Venous Catheter Systems Catalog #C1-IT. 
               Recall #Z-746-4.
CODE           Lot #05042 and 10071.
MANUFACTURER   Menlo Care, Inc., Menlo Park, California.
RECALLED BY    Manufacturer, by telephone followed by letter June 28, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   New York, Florida.
QUANTITY       Firm estimates none remains on the market.
REASON         A crack may develop in the side arm of the hemostasis valve
               with sideport component and the mainline catheter system,
               which could result in back bleeding.

_______________
PRODUCT        Promotional Materials for the Electone brand Hearing Aids. 
               Recall #Z-764-4.
CODE           All codes.
MANUFACTURER   Electone, Inc., Longwood, Florida.
RECALLED BY    Manufacturer, by letter May 28-29, 1993.  Firm-initiated
               field correction ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       197 dispensers.
REASON         Advertising and promotional materials contain misleading
               statements that could result in consumers having unrealistic
               expectations about the products benefits.

                                    -9-
                                     
END OF ENFORCEMENT REPORT FOR MAY 18, 1994.  BLANK PAGES MAY FOLLOW.
                                   ####