FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
05/11/1994

Recalls and Field Corrections:  Foods -- Class I -- 05/11/1994

May 11, 1994                                                   94-19        
                                     
RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT        (a) Satay's Vegetarian Marinade; (b) Gado-Gado Dressing, in
               1.5, 8 and 16 ounce jars.  Recall #F-392/393-4.
CODE           (a) All lots; (b) 407019, 408009, 408030, 409020, 409025,
               401022, 401023, 401106, 401113, 410009, 410001.
MANUFACTURER   Texas Food Research, Austin, Texas.
RECALLED BY    Manufacturer, by letter dated February 8, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 4 cases of 1.5 ounce jars (6 jars per case), 175 cases
               of 8 ounce jars (4 jars per case), 8 cases of 16 ounce jars
               (12 jars per case); (b) 7 cases 1.5 ounce jars (6 jars per
               case), 152 cases 8 ounce jars (4 jars per case).
REASON         Product has the potential to support the outgrowth of
               Clostridium botulinum toxin.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Rugby brand: (a) Tri-Statin II Cream (Nystatin and
               Triamcinolone Acetonide Cream, USP), in 15 gram tubes, Rx,
               for dermatologic use; (b) Triamcinolone Acetonide Cream, USP
               0.025%, in 15 gram tubes, Rx, for dermatological use.
               Recall #D-264/265-4.
CODE           Lot #3K644 EXP 11/95.
MANUFACTURER   NMC Laboratories, Inc., Glendale, New York.
RECALLED BY    Manufacturer, by telephone April 10, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   1 wholesale account in New York.
QUANTITY       8,870 tubes were distributed.
REASON         A portion of the lot of Tri-Statin II Cream (Nystatin &
               Triamcinolone Acetonide) tubes were miscartoned as
               Triamcinolone Acetonide Cream.

_______________
PRODUCT        Rx oral suspension antibiotics used for the treatment of
               respiratory infections, relabeled by Allscrips:
               (a) Ceclor Cefaclor Oral Suspension -- 
               125 mg/5 ml, in 75 ml, 150 ml bottles;
               187 mg/5 ml, in 50 ml and 100 ml bottles; 
               250 mg/5 ml, in 75 ml bottles and 150 ml bottles,
               375 mg/5 ml, in 50 ml bottles and 100 ml bottles,
               (b) Lorabid Loracarbef for Oral Suspension -- 
               100 mg/5 ml, in 50 ml bottles and 100 ml bottles; 
               200 mg/5 ml, in 50 ml bottles and 100 ml bottles; 
               (c) Keflex Cephalexin Oral Suspension, Manufactured by Dista
               Products Co., Carolina, Puerto Rico -- 
               250 mg/5 ml, in 100 ml bottles and 200 ml bottles.  
               Recall #D-267/273-4.
CODE           All unexpired lots. 
MANUFACTURER   (a&b) Eli Lilly Industries, Inc., Carolina, Puerto Rico
               (responsible firm); (c) Dista Products Company, Carolina,
               Puerto.
RECALLED BY    Allscrips Pharmaceuticals Inc., Vernon Hills, Illinois
               (relabeler), by letter April 1, 1994.  Firm-initiated recall
               ongoing.  See also FDA statement: Recall of Liquid
               Antibiotics, dated March 31, 1994.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 26,898 bottles; (b) 1,190 bottles; (c) 59 bottles were
               distributed; firm estimates that 25 percent of the product
               remains on the market.
REASON         Bottles may contain a small plastic cap.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Sterile Lubraseptic Jelly, in 10 g sterile tubes, non-
               prescription product used for lubrication of catheters,
               cystoscopes, instruments, and gloves, has anesthetic and
               antibacterial qualities, under Baker Norton and United
               Guardian labels.  Recall #D-266-4.
CODE           All lots within expiration date.  EXP dates range from 
               5/94 - 11/98.
MANUFACTURER   United Guardian, Hauppauge, New York.
RECALLED BY    Manufacturer, by letter April 12, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       43,996 cartons (24 per carton) were distributed.
REASON         Subpotency.

                                    -2-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Platelets, Pheresis;
               (d) Cryoprecipitated AHF; (e) Red Blood Cells, For Further
               Manufacture; (f) Fresh Frozen Plasma; (g) Recovered Plasma. 
               Recall #B-253/259-4.
CODE           Unit numbers:  (a) FR32930, FR33173, FR37775, FR38903,
               FR39782, GH47903, GH50280, GH57430, GK07162, GK14511,
               GM06352, GM16088, GM22426, Q63480, Q66976;
               (b) FR32930, FR33173, FR35764, FR38903, FR39782, GH50049,
               GH52410, GH54742, GH57430, GM04902, GM14926;
               (c) LG02745; (d) FR38903, GK07162, GM19997, Q66976; 
               (e) Q64223; (f) FR37775, FR39620, FR39782, GK13899, GM06352,
               Q64223; (g) FR32930, FR33173, FR35764, FR38903, GH47903,
               GH49331, GH50049, GH50280, GH51479, GH52410, GH56633,
               GH57430, GK07162, GK14511, GK16905, GK19475, GM04902,
               GM14926, GM16088, GM19997, GM22426, LG13705, LJ16704,
               Q63480, Q66976.
MANUFACTURER   Central Blood Bank, Pittsburgh, Pennsylvania.
RECALLED BY    Manufacturer, by telephone May 4 and 18, 1993, and by fax
               May 19, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania, New York, Illinois, North Carolina,
               California.
QUANTITY       (a) 15 units; (b) 11 units; (c) 1 unit; (d) 4 units; (e) 1
               unit; (f) 6 units; (g) 25 units.
REASON         Blood products, which tested negative for the antibody to
               human immunodeficiency virus type 1 (anti-HIV-1), but were
               collected from a donor who previously tested repeatedly
               reactive for anti-HIV-1, were distributed.

_______________
PRODUCT        Red Blood Cells.  Recall B-260-4.
CODE           Unit #12LH06407.
MANUFACTURER   American Red Cross Blood Services, Charlotte, North
               Carolina.
RECALLED BY    Manufacturer, by telephone September 14, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   North Carolina.
QUANTITY       1 unit.
REASON         Blood product, maintained at improper temperatures during
               shipment, was distributed.

_______________
PRODUCT        (a) Platelets, Pheresis.  Recall #B-267-4.
CODE           Unit #C26551.
MANUFACTURER   Northern Illinois Blood Bank, Inc., Rockford, Illinois.
RECALLED BY    Manufacturer, by telephone October 29, 1993, followed by
               letter November 23, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit.

                                    -3-REASON         Blood product, which tested initially reactive for hepatitis
               B surface antigen (HBsAg) followed by a single repeat test,
               was distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        (a) Platelet Concentrate; (b) Platelets, Pheresis.
               Recall #B-250/251-4.
CODE           Unit numbers:  (a) C19081, C19083, C19109, C19113, C19115,
               C19120, C19121, C19122, C19123, C19125, C19126, C19128,
               C19129, C19131, C19132, C19135, C19136, C19138, C19139,
               C19142, C19143, C19144, C19145, C19149, C19151, C19155,
               C19156, C19157, C19159, C19160, C19161, C19162, C19163,
               C19165, C19166, C19167, C19169, C19171, C19172, C19173,
               C19174, C19175, C19177, C19178, C19179, C19180, C19181,
               C19183, C19185, C19187, C19188, C19189, C19190, C19191,
               C19192, C19193, C19195, C19196, C19198, C19199, C19200,
               C19201, C19203, C19204, C19209, C19210, C19211, C19212,
               C19213, C19214, C19216, C19219, C19220, C19221, C19223,
               C19224, C19225, C19227, C19228, C19230, C19231, C19232,
               C19233, C19234, C19235, C19236, C19237, C19238, C19240,
               C19241, C19243, C19244, C19247, C19249, C19252, C19253,
               C19254, C19255, C19258, C19260, C19261, C19265, C19267,
               C19268, C19272, C19273, C19274, C19275, C19276, C19278,
               C19279, C19280, C19283, C19285, C19286, C19287, C19289,
               C19290, C19291, C19292, C19293, C19294, C19295, C19297,
               C19299, C19300, C19303, C19304, C19305, C19313, C19316,
               C19317, C19318, C19319, C19320, C19321, C19322, C19324,
               C19325, C19326, C19328, C19329, C19330, C19331, C19333,
               C19336, C19342, C19564, C19570, C19573, C19581, C19583,
               C19584, C19588, C19589, C19640, C19648, C19650, C19652,
               C19658, C19659, C19667, C19670, C19673, C19675, C19741,
               C19748, C19756, C19768, C19777, C19778, C19783, C19787,
               C19788, C19790, C19800, C19802, C19807, C19809, C19810,
               C19811, C19814, C19815, C19816, C19818, C19820, C19821,
               C19823, C19824, C19826, C19827, C19829, C19830, C19831,
               C19832, C19833, C19835, C19836, C19837, C19838, C19845,
               C19871, C19877, C19891, C19896, C19922, C19955, C19963,
               C19965, C19970, C19971, C19972, C19973, C19977, C19979,
               C19982, C20482, C20484, C20489, C20490, C20496, C20500,
               C20510, C20513, C20514, C20518, C20524, C20526, C20528,
               C20529, C20531, C20532, C20533, C20534, C20538, C20540,
               C20541, C20542, C20545, C20547, C20548, C20550, C20552,
               C20554, C20555, C20556, C20557, C20558, C20559, C20560,
               C20561, C20564, C20566, C20567, C20571, C20576, C20578,
               C20579, C20581, C20587, C20589, C20590, C20591, C20592,
               C20593, C20594, C20596, C20597, C20599, C20600, C20601,
               C20602, C20604, C20605, C20606, C20607, C20608, C20610,
               C20612, C20613, C20615, C20621, C20622, C20623, C20624,

                                    -4-               C20625, C20626, C20627, C20628, C20630, C20634, C20637,
               C20646, C20652, C20656, C20666, C20668, C20669, C20671,
               C20672, C20676, L33339, L33343, L33344, L33348, L33351,
               L33355, L33362, L33368, L33374.
               (b) C19086, C19112, C19184, C19215, C19217, C19239, C19251, 
               C19259, C19284, C19288, C19298, C19309, C19310, C19312,
               C19643, C19647, C19742, C19743, C19745, C19785, C19803,
               C19804, C19805, C19808, C19822, C19825, C19905, C20471,
               C20504, C20508, C20546, C20549, C20563, C20582, C20583,
               C20584, C20614, C20616, C20618, C20619, C20673.
MANUFACTURER   Northern Illinois Blood Bank, Rockford, Illinois.
RECALLED BY    Manufacturer, by letters of August 16, 1993, and September
               16, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       (a) 310 units; (b) 41 units.
REASON         Platelets, stored under unacceptable temperatures, were
               distributed.

_______________
PRODUCT        Recovered Plasma.  Recall #B-261-4.
CODE           Unit numbers 12K80662, 12Y46475.
MANUFACTURER   American Red Cross Blood Services, Charlotte, North
               Carolina.
RECALLED BY    Manufacturer, by Fax September 27, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   California.
QUANTITY       2 units.
REASON         Blood product collected from a donor with a history of
               cancer, was distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
               Recall #B-262/264-4.
CODE           Unit numbers:  49L59671, 49L59672.
MANUFACTURER   American Red Cross Blood Services, Tulsa, Oklahoma.
RECALLED BY    Manufacturer, by telephone June 30, 1993, July 8, 1993, and
               by letter July 26, 1993, and by Fax July 20, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Oklahoma, California.
QUANTITY       2 units of each component.
REASON         Blood products manufactured using an unacceptable collection
               and processing technique, were distributed.

_______________
PRODUCT        (a) Whole Blood CPDA-1; (b) Red Blood Cells.
               Recall #B-265/266-4.
CODE           Unit numbers:  (a) C26516; (b) C26528, C26547, C26563.
MANUFACTURER   Northern Illinois Blood Bank, Inc., Rockford, Illinois.
RECALLED BY    Manufacturer, by telephone October 29, 1993, followed by
               letter November 23, 1993.  Firm-initiated recall complete.

                                    -5-DISTRIBUTION   Illinois, Wisconsin.
QUANTITY       (a) 1 unit; (b) 3 units.
REASON         Blood products, which tested initially reactive for
               hepatitis B surface antigen (HBsAg) followed by a single
               repeat test, were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS I
========================
_______________
UPDATE         Recall #Z-299/300-4, which appeared in the May 4, 1994,
               Enforcement Report should be corrected as follows:
               PRODUCT: Defib-Pads, catalog numbers 2345, 2345N, 2346, and
                        2346 N.
               CODE:    Manufacturing dates of JUN.92 through JAN.94 on the 
                        product labels and corresponding "Use By" dates of
                        JUN.95 through Jan.97.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Evershears Bipolar Coagulating Scissors, Model 3801, used
               during laparoscopic surgical procedures for mechanical
               cutting and bipolar coagulation using electrosurgical
               energy.  Recall #Z-676-4.
CODE           Lot numbers:  XJ934301, AC004326, XJ934601.
MANUFACTURER   Everest Medical Corporation, Minneapolis, Minnesota.
RECALLED BY    Manufacturer, by telephone February 28, 1994, and by letter
               March 2, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, United Kingdom.
QUANTITY       175 units from lots XJ934301 and AC004326 and 3 units from
               lot XJ934601 were distributed.  
REASON         The ceramic blade of the scissor separates from the metal
               support blade during use.

_______________
PRODUCT        (a) PCA-10 3 Ring with Rotator 10 cc Disposable Angiography
               Control Syringe, sterilized with ethylene oxide gas, non-
               pyrogenic, Catalog #C305-0035; (b) Coeur PCA-10 3 Ring with
               Rotator 10 cc Disposable Angiograpy Control Syringe, bulk,
               non-sterile, Catalog #C305-1035, used for injections into a
               patient's vascular system.   Recall #Z-703/704-4.
CODE           Lot numbers:  (a) 50793701, 50893701, 50893702; (b)
               50793801, 50893801, 50993801, 51093801, 50593801, 50693801.
MANUFACTURER   Coeur Laboratories, Inc., Raleigh, North Carolina.
RECALLED BY    Manufacturer, by telephone and letter on December 6, 1993. 
               Firm-initiated recall complete.
DISTRIBUTION   California, New Jersey, Florida, Kentucky, Texas, Virginia,
               England, Germany, Australia, Canada.
QUANTITY       (a) 1,600 units; (b) 10,900 units were distributed.

                                    -6-REASON         Product defect which could allow air bubbles to leak into
               the syringe and be injected into the patient.

_______________
PRODUCT        TMJ Abrasion Burs for small joint arthroscopy sold under the
               Stryker Endoscopy label:
               (a) Model 266-641-000, 2 mm; (b) Model 266-647-000, 3 mm.
               Recall #Z-709/710-4.
CODE           Lot numbers:  (a) W0009476-1; (b) W0009477-1.
MANUFACTURER   Stryker Puerto Rico, Inc., Arroyo, Puerto Rico.
RECALLED BY    Stryker Endoscopy, San Jose, California, by letter March 25,
               1993.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Hong Kong, West Germany, Mexico.
QUANTITY       (a) 54 units; (b) 30 boxes.
REASON         The burs did not have suction slots cut into them.

_______________
PRODUCT        Medtronic Elite II, Model 7084 Activity Responsive Dual
               Chamber Pacemakers.  Recall #Z-720-4.
CODE           Serial numbers: 3U2000156H-3U2000162H, 3U2000165H-
               3U2000168H, 3U2000181H-3U2000184H, 3U2000199H, 3U2000200H,
               3U2000202H-3U2000204H, 3U2000279H, 3U2000280H, 3U2000283H -
               3U2000286H, 3U2000289H, 3U2000291H-3U2000293H, 3U2000295H -
               3U2000299H, 3U2000301H - 3U2000303H, 3U2000305H-3U2000307H,
               3U2000311H, 3U2000317H, 3U2000321H, 3U2000323H, 3U2000326H,
               3U2000327H, 3U2000331H, 3U2000334H, 3U2000340H, 3U2000341H,
               3U2000349H-3U2000367H.  
MANUFACTURER   Med Rel, Inc., Humacao, Puerto Rico.
RECALLED BY    Medtronic, Inc., Minneapolis, Minnesota, by telephone March
               10, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Alaska, California, Colorado, Florida, Georgia, Hawaii,
               Iowa, Maryland, Minnesota, New Hampshire, New York, North
               Carolina, Oklahoma, Pennsylvania, South Carolina, Tennessee,
               Texas.
QUANTITY       70 units were distributed; firm estimates none remains on
               the market.
REASON         The devices were distributed into interstate commerce
               without an approved Premarket Approval supplement.

_______________
PRODUCT        Pal Pro-Assist Professional Assistance Lift and Full Body
               Transfer Lift, motorized patient transport devices: 
               (a) Model PA100; (b) Model PA200; (c) PA300; 
               (d) Model PL150; (e) Model PL250.  Recall #Z-721/725-4.
CODE           All units.
MANUFACTURER   Sunrise Machine & Tool, Inc., Detroit Lakes, Minnesota.
RECALLED BY    Manufacturer, by letter December 21, 1993.  Firm-initiated
               field correction complete.
DISTRIBUTION   Colorado, Georgia, Iowa, Kentucky, Massachusetts, Michigan,
               Minnesota, Montana, North Dakota, South Dakota, Oregon,
               Pennsylvania, Tennessee, Texas, Wisconsin.
                                    -7-QUANTITY       124 units.
REASON         The legs of the patients lifts could come out of their
               channels when used in the narrow position, allowing the main
               mast to fall forward, dropping the patient to the floor.

_______________
PRODUCT        Night Vision Image Intensifiers, Model T3C-2, used for
               viewing objects and persons in subdued lighting.  
               Recall #Z-748-4.
CODE           Model T3C-2.
MANUFACTURER   Novosibirsk Instrument Making Plant, in Novosibirsk Russian
               Federation.
RECALLED BY    Sentinel Electronics, Bristol, Pennsylvania (importer).  FDA
               approved the firm's corrective action plan April 18, 1994. 
               Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       16 units.
REASON         Product contains a radiation defect, in that the products
               emit x-radiation that is unnecessary to the accomplishment
               of the purpose of the product, and create a risk of injury
               to users.

_______________
PRODUCT        Arrow Two-Lumen Central Venous Catherization Kit containing
               the spring wire guide, which was the subject of the
               packaging error.  The spring wire guide is used to guide the
               catheter into the correct position.  Recall #Z-753-4.
CODE           Lot #AK-2-260-3.
MANUFACTURER   Arrow International, Inc., Randleman, North Carolina.
RECALLED BY    Arrow International, Inc., Reading, Pennsylvania, by visit
               starting October 11, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Georgia, New Jersey, California, New York.
QUANTITY       30 cases (5 kits per case) were distributed; firm estimates
               none remains on the market.
REASON         The catheter is not usable with a 0.035" diameter spring
               wire guide (which was incorrectly packaged into the recalled
               catheter kit).  

_______________
PRODUCT        HAID Universal Bone Plate (UBP) System, used for a variety
               of operative procedures:
               (a) Part #650005-002, Catalog No. S/L 50-4551 (Small Set); 
               (b) Part #650005-003, Catalog No. S/L 50-4552 (Standard
               Set); 
               (c) Part #650005-004, Catalog No. S/L 50-4553 (Large Set); 
               (d) Part #650005-005, Catalog No. S/L 50-4554 
               (Evaluation Set).   Recall #Z-755/758-4.
CODE           All units manufactured and distributed between July 16, 1991
               and February 9, 1993.
MANUFACTURER   American Medical Electronics, Inc. (AME), Richardson, Texas.

                                    -8-RECALLED BY    Manufacturer, by letter February 9, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       548 UBP systems and an undetermined quantity of bone plates
               and screws ordered as refills.
REASON         The new intended use for the devices in the cervical spine
               (fixation/stabilization) does not have a required premarket
               approval.   


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        Sterile Gelfilm absorbable gelatin film, USP and Sterile
               Ophthalmic Gelfilm, USP, for use in neurosurgery and
               thoracic and ocular surgery.  Recall #Z-632-4.
CODE           Lot numbers:  407FR, 438FS, 414FR, 442FS.
MANUFACTURER   The Upjohn Company, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter April 15, 1993.   Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       891 units were distributed.
REASON         Products failed to meet the USP digestibility specification
               within the labeled expiration period.

_______________
PRODUCT        Sterile 7 x 30 mm Cannu-Flex Silk Interference Screw,
               Catalog #014202, in double tyvek blister package.  
               Recall #Z-685-4.
CODE           Lot #001097.
MANUFACTURER   Acufex Microsurgical, Inc., Mansfield, Massachusetts.
RECALLED BY    Manufacturer, by voice mail January 5, 1994, and by letter. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       106 units were distributed.
REASON         Product was labeled 7 x 30 mm CannuFlex SilkSrew, but
               actually contained non-cannulated 7 x 30 mm Interference
               Screw.

_______________
PRODUCT        Abbott IMx CK-MB Controls, for verification of the
               calibration of the IMx Analyzer used for the determination
               of human CK-MB, in 3 - 8 ml bottles; (b) Abbott IMx CK-MB
               Mode 1 Calibrator, for in-vitro diagnostic use with IMx CK-
               MB, in 3 - 4 ml bottles with (human) heart CK-MB in human
               serum with protein stabilizers, 30 mg/ml.  
               Recall #Z-730/731-4.
CODE           Lot numbers:  (a) 78833Q100 EXP 3/18/94, 80198M200 EXP
               5/9/94; (b) 80627M200 EXP 5/9/94.
MANUFACTURER   Abbott Laboratories, Diagnostics Division Abbott Park,
               Illinois.
                                    -9-RECALLED BY    Manufacturer, by telephone beginning August 31, 1993.  Firm-
               initiated recall complete.
DISTRIBUTION   Nationwide, Canada, France, Japan, Korea, Singapore.
QUANTITY       774 control packs and 604 Mode 1 calibrators were
               distributed.
REASON         Devices were found to have loss potency due to thermal
               stress causing chemical degradation, which may result in
               controls reading outside of specifications causing invalid
               assay results.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
==========
_______________
PRODUCT        Urea bolus, 13.4 grams/bolus, packaged in 50 and 25 pound
               bottles, an OTC intrauterine antiseptic distributed under
               the following labels:
               (a) Antisepsis Bolus, manufactured by Sparhawk Laboratories,
               Inc., Kansas City, KS;
               (b RXV Products, After-Birth, manufactured for RX Veterinary
               Products, Porterville, CA 
               (c) Uterine Bolus, manufactured for Animal Health Associates
               Kansas City, MO 
               (d) AGRIpharm After-Birth, manufactured for Dealer          
               Distribution of America, Kansas City, MO 
               (e) Butler Uterine Bolus, manufactured for The Butler
               Company, Columbus, OH 
               (f) AgriLabs Intrauterine Bolus, manufactured for Agri
               Laboratories, Ltd., St. Joseph, MO 
               (g) Uterine Bolus, manufactured for Phoenix Pharmaceutical
               Inc., St. Joseph, MO 
               (h) Animal Medic Uterine Medic Bolus, manufactured for 
               Animal Medic Inc., Manchester, PA 
               (i) L and M Pharmaceuticals Intrauterine Bolus, manufactured
               for L and M Pharmaceuticals, Dallas, Tx.  Recall #V-042-4.
CODES          (a) 930307 through 930310; (b) 930312; (c) 930321;
               (d) 930323; (e) 930406 through 930411; (f) 930455, 930456 
               (g) 930463 through 930465; (h) 930512 through 930515;
               (i) 930536 through 930539.
MANUFACTURER   Veterinary Laboratories, Inc., Lenexa, Kansas.
RECALLED BY    Manufacturer, by telephone on or about June 15, 1993, and by
               letter March 18, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Missouri, Alabama, Colorado, Nebraska, California, New York,
               Iowa, Maine, Pennsylvania, Ohio, Texas, Massachusetts 
               and Puerto Rico.
QUANTITY       3,803 50-bolus bottles and 300 25-bolus bottles of lots
               930307 through 930323 were distributed.
               13,117 50-bolus bottles and 300 25-bolus bottles of lots
               930406 through 930539 were distributed.

                                   -10-REASON         Lots 930307 through 930323 recalled due to stability samples
               swelling and becoming soft.  Lots 930406 through 930539 were
               subpotent. 


SEIZURES:
=================================================================
_______________
PRODUCT        Picked Crab Meat, Cooked Whole Crabs, and Crab Shells
               (intended for use as serving dishes)(94-682-795/809).
CHARGE         Adulteration - One of the articles bears or contains a 
               poisonous and deleterious substance, the microorganism
               Listeria monocytogenes, which may render it injurious to
               health; some of the articles consist in part of a filthy
               substance by reason of the presence therein of animal hair;
               some of the articles consist in part of a filthy substance
               by reason of the presence therein of the bacteria
               Escherichia coli; some of the articles consist in part of a
               decomposed substance by reason of the presence therein of
               decomposed crab meat; and all of the articles have been
               prepared, packed, and held under insanitary conditions
               whereby they may have become contaminated with filth. 
LOCATION       Barwick Crab Company, Crawfordville, Florida. 
FILED          March 29, 1994; U.S. District Court for the Northern
               District of Florida, Tallahassee Division; Civil #CIV-94-
               40130; FDC #66931.
SEIZED         April 19, 1994 - goods valued at approximately $15,000.

_______________
PRODUCT        Frozen Yellow Fin Tuna Loins (94-447-479).
CHARGE         Adulterated - The article consists in part of decomposed
               fish.
LOCATION       Alterman Cold Storage Company, Pensacola, Florida.
FILED          March 16, 1994; U.S. District Court for the Northern
               District of Florida, Pensacola Division; Civil
               #CIV-94-30091; FDC #66942.
SEIZED         March 21, 1994 - goods valued at approximately $12,000.

_______________
PRODUCT        Frozen Yellow Fin Tuna Steaks (94-629-602).
CHARGE         Adulterated - The article consists in part of decomposed 
               fish. 
LOCATION       Cityice Cold Storage Company, LaConner, Washington.
FILED          March 22, 1994; U.S. District Court for the Western 
               District of Washington; Civil #C94-0415Z; FDC #66947.
SEIZED         March 31, 1994 - goods valued at approximately $35,000.

_______________
PRODUCT        Sodium Citrate (94-688-047).
CHARGE         Adulterated - The article is unfit for food because of the 
               presence of foreign material including plant matter and a
               rubber-like material visible to the naked eye.
                                   -11-LOCATION       G.S. Robins and Company, Springfield, Missouri.
FILED          April 12, 1994; U.S. District Court for the Western
               District of Missouri, Southern Division; Civil #94-3152-CV-
               S-4; FDC #66954.
SEIZED         April 29, 1994 - goods valued at approximately $19,127.

                                   -12-
                                   ####


END OF ENFORCEMENT REPORT FOR MAY 11, 1994.  BLANK PAGES MAY FOLLOW.