FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
04/06/1994

Recalls and Field Corrections:  Drugs -- Class II -- 04/06/1994

APRIL 6, 1994                                                   94-14

RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Food Industry Sanitizing Soap (GOJO), distributed under ten
               (10) private labels and product names in addition to the
               firm's GOJO brand; and is packaged in five (5) different
               sizes as follows:  
               1.  GOJO brand E-2 Food Industry Sanitizing Soap, in 800 ml
               cartridges, 1000 ml cartridges, 2000 ml cartridges, 5000 ml
               cartridges, and in 4 ounce bottles; 
               2.  Easterday Janitorial Supply brand E-2 Germicidal Soap
               Plus, in 1000 ml cartridges, 6 units per case;
               3.  Hillyard Chemical brand E-2 Plus Food Industry 
               Sanitizing Soap, in 4 ounce bottles, 12 units per case; 1000
               ml cartridges, 6 units per case; 2000 ml cartridges, 4 units
               per case;
               4.  Peninsular Paper brand E-2 1000 Food Industry Sanitizing
               Soap in 1000 ml cartridges,  6 units per case;
               5.  North American Paper brand Spectrum E-2 Food Industry
               Sanitizing Soap, in 1000 ml cartridges, 6 units per case;
               6.  Penn Jersey brand Sanitary Supply Sani-Soap Food
               Industry Sanitizing Soap, in 1000 ml cartridges, 6 units per
               case;
               7.  Massco brand Food Prep Sanitizing Soap, in 1000 ml
               cartridges, 6 units per case; 
               8.  SourceOne brand E-2 Food Industry Sanitizing Soap, in
               1000 ml cartridges, 6 units per case;
               9.  International Distribution System (IDS), brand E-2 Food
               Industry Sanitizing Soap, in 1000 ml cartridges, 6 units per
               case, 5000 ml cartridges, 6 units per case;
               10.  Selig Chemical Industries brand Sootho Food Processing  
               Sanitizing Lotion Soap E-2, in 800 ml cartridge, 12 units
               per case; 
               11.  Zep Manufacturing brand, Handstand TM E-2 Hand Soap, in 
               800 ml cartridges, 12 units per case.  Recall #D-221-4.
CODE           Batch numbers:  91537, 92357, 05475, and 12778.
MANUFACTURER   GOJO Industries, Inc., Cuyahoga Falls, Ohio. 
RECALLED BY    Manufacturer, by letter March 15, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       4,100 cases were distributed.
REASON         Presence of Pseudomonas cepacia and Serratia marcescens.

_______________
PRODUCT        Various Rx injectable drug products, 25 vials per tray:
               (a) Aminophylline Injection, USP, 25 mg/ml, 100 ml vials,
               pharmacy bulk pack; for IV use only after dilution for the
               relief of acute bronchial asthma;
               (b) Clindamycin Phosphate Injection, USP, 150 mg/ml, a
               semisynthetic antibiotic for IM or IV use for the treatment
               of serious infections caused by susceptible anaerobic
               bacteria; 2 ml vials, 4 ml vials, 6 ml vials;
               (c) Glycopyrrolate Injection, USP, 0.2 mg/ml, an
               anticholinergic for the use as a preoperative antimuscarinic
               to reduce salivary, tracheobronchial and pharyngeal
               secretions, 1 ml vials, 2 ml vials, 5 ml vials, 20 ml vials;
               (d) Heparin Sodium Injection, USP, for anticoagulant therapy
               in prophylaxis and treatment of venous thrombosis and the
               prevention of postoperative deep venous thrombosis and
               pulmonary embolism:  1000 USP units/ml 10 ml vial, 30 ml
               vials, (private label) - Rx/Dx Hemochron Heparin Sodium
               Injection, USP, 1,000 USP units/ml, 10 ml multiple dose
               vial, Manufactured for International Technidyne Corp., 
               Edison, NJ;  10,000 USP units/ml 5 ml vials; 
               (e) Heparin Lock Flush Solution, USP, 10 units/ml, for the
               maintenance of patency of an indwelling venipuncture device
               designed for intermittent injection of infusion therapy or
               blood sampling 10 ml vial, 30 ml vials;
               (f) MVC 9 + 3 Multivitamin Concentrate, a daily multivitamin
               dosage for adults and children aged 11 and above receiving
               parenteral nutrition (for infusion after dilution) 10 ml
               covial, 50 ml vials;
               (g) Potassium Chloride Injection Concentrate, USP, 2 mEq/ml,
               IV infusion after dilution for the treatment potassium
               deficiency states when oral replacement is not feasible; 100
               ml vial, 200 ml vials;
                                    -2-               (h) Sterile Cefazolin Sodium, USP, a semisynthetic
               cephalosporin antibiotic for parenteral administration for
               the treatment of respiratory tract infections, genital tract
               infections and other serious infections due to susceptible
               organisms:  500 mg, 10 ml vial, 100 ml vial;  1 g, 10 ml
               vial, 100 ml vial;  10 g, 100 ml vial;
               (i) Vincristine Sulfate Injection, USP, 1 mg/ml, used for
               the treatment of acute leukemia; 1 ml vial; 2 ml vial,       
               5 ml vial.  Recall #D-222/233-4.
CODE           All lots within expiration date.   (Note: Cefazolin Sodium
               products with lot numbers ending in "A" are not subject to
               recall). 
MANUFACTURER   Lyphomed/Fujisawa USA, Inc., Melrose Park, Illinois 
               (products a, b, c, f, g, i). 
               Lyphomed/Fujisawa USA, Inc., Grand Island, New York
               (products d, e, h).
RECALLED BY    Fujisawa USA Deerfield, Illinois, by letter March 7, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Aminophylline - Nationwide; Clindamycin - Nationwide, Malta,
               Qatar; Glycopyrrolate - Nationwide; Heparin - Nationwide,
               Israel, Chile, Mexico, Palau, and the Bahamas;
               MVC 9+3 - Nationwide and international;
               Potassium Chloride - Nationwide and Israel;
               Cefazolin - Nationwide and New Zealand;
               Vincristine - Nationwide and Belize.
QUANTITY       5,400 vials of aminophylline, 50,600 vials of clindamycin,
               3,833,220 vials of glycopyrrolate, 3,575,075 vials of 1,000
               unit heparin, 1,011,775 vials of 10,000 unit heparin,
               452,300 vials of heparin lock flush, 818,325 vials of MVC
               9+3, 520,700 vials of potassium, 197,200 vials of 500 mg
               cefazolin, 743,990 vials of 1 g cefazolin, 27,450 vials of
               10 g cefazolin, and 56,544 vials of vincristine were
               distributed, with the firm estimating that 15-20% of the    
               product remains on the market. 
REASON         Abbreviated New Drug Application/Abbreviated Antibiotic
               Ddrug Application discrepancies.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Bisacodyl Tablets, 5 mg, nonprescription laxative.  
               Recall #D-220-4.
CODE           54073B and 53073B.
MANUFACTURER   Vintage Pharmaceuticals, Inc., Charlotte, North Carolina.
RECALLED BY    Manufacturer, by telephone and by letter February 6, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Minnesota, New Jersey.
QUANTITY       485,697 tablets of lot 54073B and 323,798 tablets of lot
               53073B were distributed.
REASON         Use of non-food grade ink on tablets.  Ink manufacturer
               responsible for error.

                                    -3-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II ---------------------
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-165/166-4.
CODE           Unit numbers: (a) 04G43485, 04G43488, 04G43491, 04G43492,
               04G43493, 04G43495, 04G43497, 04G43499, 04G43501, 04G43502,
               04G43503, 04G43504, 04G43505, 04G43506, 04G43507, 04G43508,
               04G43511, 04G43512, 04G43513, 04G43514, 04G43515, 04G43516,
               04G43518, 04G43519, 04G43521, 04G43523, 04H91883, 04H91885,
               04H91886, 04H91887, 04H91891, 04H91892, 04H91893, 04H91894,
               04H91895, 04H91896, 04H91898, 04H91899, 04H91901, 04H91903,
               04H91906, 04H91908, 04H91909, 04H91910, 04H91914, 04H91916,
               04H91918, 04H91920, 04H91922, 04H91924, 04H91925, 04H91926,
               04H91927, 04H91928, 04H91930, 04H91931, 04H91932, 04H91933,
               04H91936, 04H91940, 04H91941, 04H91942, 04H91946, 04H91947,
               04H91949, 04H91950, 04H91951, 04H91952, 04H91953, 04H91954,
               04H91957, 04H91958, 04H91960, 04H91961, 04H91962, 04H91963,
               04H91964, 04H91965, 04GF18083, 04GF18084, 04GF18085,
               04GF18087, 04GF18089, 04GF18090, 04GF18091, 04GF18092,
               04GF18095, 04GF18096, 04GF18097, 04GF18101, 04GF18104,
               04GF18105, 04GF18106, 04GF18108, 04GF18109, 04GF18110,
               04GF18113, 04GF18114, 04GF18117, 04GF18120, 04GF18122,
               04GF18123, 04GF18125, 04GF18127, 04GF18128, 04GF18130,
               04GF18135, 04GF18137;
               (b) 04G43485, 04G43488, 04G43489, 04G43491, 04G43492,
               04G43493, 04G43494, 04G43495, 04G43497, 04G43499, 04G43501,
               04G43511, 04G43512, 04G43513, 04G43515, 04G43516, 04G43517,
               04G43518, 04G43519, 04G43521, 04G43522, 04G43523, 04H91886,
               04H91888, 04H91891, 04H91892, 04H91893, 04H91894, 04H91895,
               04H91896, 04H91897, 04H91902, 04H91903, 04H91904, 04H91905,
               04H91906, 04H91907, 04H91908, 04H91909, 04H91911, 04H91912,
               04H91913, 04H91915, 04H91916, 04H91917, 04H91918, 04H91920,
               04H91922, 04H91924, 04H91928, 04H91930, 04H91931, 04H91933,
               04H91936, 04H91939, 04H91941, 04H91942, 04H91944, 04H91945,
               04H91946, 04H91947, 04H91949, 04H91950, 04H91952, 04H91953,
               04H91954, 04H91956, 04H91959, 04H91960, 04H91963, 04H91965,
               04H91966, 04GF18081, 04GF18083, 04GF18086, 04GF18087,
               04GF18088, 04GF18090, 04GF18093, 04GF18099, 04GF18100,
               04GF18102, 04GF18104, 04GF18106, 04GF18107, 04GF18108,
               04GF18110, 04GF18111, 04GF18112, 04GF18116, 04GF18117,
               04GF18118, 04GF18119, 04GF18121, 04GF18125, 04GF18127,
               04GF18128, 04GF18130, 04GF18131, 04GF18133, 04GF18134,
               04GF18136. 
MANUFACTURER   American Red Cross Blood Services, Dedham, Massachusetts.
RECALLED BY    Manufacturer, by telephone July 10, and 15, 1991, and by
               letter July 24, 1991.  Firm-initiated recall complete.
DISTRIBUTION   Massachusetts, Maine.
QUANTITY       (a) 108 units; (b) 102 units.
REASON         Blood products incorrectly tested for syphilis were
               distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-188-4.
CODE           Unit #28027-4218.
                                    -4-MANUFACTURER   United Blood Services Blood Systems, Inc., San Angelo,
               Texas.
RECALLED BY    Manufacturer, by letter March 31, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Red Blood Cells, collected from a donor who had visited an
               area endemic for malaria and had taken prophylactic anti-
               malarial medication two years prior to donation, were
               distributed.

_______________
UPDATE         The February 9th Enforcement Report identified Wheaton
               Tubing Products ("Wheaton") as the manufacturer of recalled
               sterile glass vials for use with diluents and allergenic
               extracts (Recall #B-140/141-4).  Although the glass vials, a
               component of the final product, were supplied by Wheaton,
               the manufacturer is Allergy Laboratories, Inc.  The
               Enforcement should have identified Allergy Laboratories,
               Inc. as both the recalling and manufacturing firm of the
               final distributed product.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.  Recall #B-167/168-4.
CODE           Unit numbers: (a) 04C05639, 04C05640, 04C05643, 04C05644,
               04C05649, 04C05652, 04C05653, 04C05654, 04C05655, 04C05657,
               04J21634, 04N84567, 04N84568, 04N84572, 04N84576, 04N84578,
               04R64824, 04R64833, 04W68142, 04FS12369, 04FS12372,
               04FS12373, 04FS12375, 04FS12379, 04FS12381, 04FS12382,
               04FS12385, 04FS12392, 04FS12397, 04FS12407, 04FS12413,
               04FS12414, 04FS12415;
               (b) 04J21629, 04J21630, 04J21631, 04J21633, 04J21634,
               04M01140, 04M01141, 04M01143, 04M01144, 04M01145, 04M01146,
               04M01152, 04M01155, 04M01156, 04M01158, 04M01161, 04M01164,
               04N84567, 04N84568, 04N84572, 04N84573, 04N84578, 04N84581,
               04R64795, 04R64800, 04R64801, 04R64806, 04R64814, 04R64815,
               04R64817, 04R64822, 04R64826, 04R64828, 04R64829, 04R64830,
               04R64831. 
MANUFACTURER   American Red Cross Blood Services, Dedham, Massachusetts.
RECALLED BY    Manufacturer, by telephone July 10 and 15, 1991, followed by
               letter July 24, 1991.  Firm-initiated recall complete.
DISTRIBUTION   Massachusetts, Maine.
QUANTITY       (a) 33 units; (b) 36 units.
REASON         Blood products incorrectly tested for syphilis were
               distributed. 

                                    -5-_______________
PRODUCT        Source Plasma.  Recall #B-185-4.
CODE           Unit numbers:  C05877-109 through C05896-109, C05898-109
               through C05908-109, C05910-109 through C05912-109, 
               G26345-109 through G26369-109, G26371-109 through 
               G26399-109, G26401-109 through G26417-109, 
               G26419-109 through G26421-109, G26423-109 through 
               G26463-109, G26465-109 through G26531-109, 
               G26533-109 through G26536-109, G26538-109 through 
               G26547-109, G26549-109 through G26572-109, 
               G26574-109 through G26621-109, G26624-109 through 
               G26682-109, G26684-109 through G26711-109, 
               K05023-109 through K05028-109.
MANUFACTURER   San Juan Plasma, Inc., Farmington, New Mexico.
RECALLED BY    Manufacturer, by letter August 17, 1993, and by telephone
               March 7, 1994.  Firm-initiated recall complete.
DISTRIBUTION   California, Massachusetts.
QUANTITY       395 units.
REASON         Source Plasma, exposed to elevated storage temperatures and
               not properly relabeled as Source Plasma, Salvaged, was
               distributed.

_______________
PRODUCT        Platelets.  Recall #B-186-4.
CODE           Unit #C24894.
MANUFACTURER   Northern Illinois Blood Bank, Rockford, Illinois.
RECALLED BY    Manufacturer, by telephone September 27, 1993, followed by
               letter September 29, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit.
REASON         Blood product, which tested initially reactive for hepatitis
               B surface antigen (HBsAg) by EIA, was distributed.

_______________
PRODUCT        Red Blood Cells.  Recall #B-187-4.
CODE           Unit #1M99786.
MANUFACTURER   Community Blood Center of Mercer County, Trenton, New
               Jersey.
RECALLED BY    Manufacturer, by telephone June 1, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   New Jersey.
QUANTITY       1 unit.
REASON         Blood product, which was labeled with the incorrect
               expiration date, was distributed.

_______________
PRODUCT        (a) Whole Blood; (b) Platelets, Pheresis; (c) Red Blood
               Cells; (d) Platelets; (e) Recovered Plasma.
               Recall #B-189/193-4.
CODE           Unit numbers:  (a) 1075092; (b) 2604250; (c-e) 8100569.
MANUFACTURER   LifeSource, Glenview, Illinois.

                                    -6-RECALLED BY    Manufacturer, by letters of January 17 and 18, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   Illinois, New York.
QUANTITY       1 unit of each component.
REASON         Blood products which tested reactive to serologic test for
               syphilis were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        I.V. Administration Sets:  (a) "Mini Set", Order #28-60-125,
               designed to connect to the extension set used with the
               Siemens Infusion System, contained in a cardboard box, 20
               per box;
               (b) "Uni Set", Order #28-60-265, used with the Siemens
               Infusion System.  The product has two Y-sites with a check
               valve, it's priming volume is approximately 16 ml, packaged
               in a plastic pouch, contained in a cardboard box, 70 per
               box;
               (c) "Custom Set", Order #25-25-01, same as "Uni Set" with
               the exception that this product contains a filter which
               renders the custom name.  In addition, the priming volume is
               approximately 26 ml.  Recall #Z-787/789-3.
CODE           Lot numbers:  (a) Q1470 AND Q1501; (b) Q1720; 
               (c) Q1721, Q1729, Q1733, Q1737. 
MANUFACTURER   Siemens Infusion Systems, Sylmar, California.
RECALLED BY    Manufacturer, by letters of April 20, 1993, May 12 and 19,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   (a) California, Illinois, Texas, Virginia, Florida, North
               Carolina, New Jersey, Virginia, Illinois; (b) Illinois; (c)
               Illinois, Kentucky, Missouri, Mississippi, North Carolina,
               New Jersey, Texas, Virginia, West Virginia, California,
               Hawaii, Minnesota, Montana, Oklahoma, Tennessee, Wisconsin,
               Indiana.
QUANTITY       (a) 1,800 units of lot Q1470, 6,120 units of lot Q1501; (b)
               840 units of lot Q1720; (c) 2,170 units of lot Q1721, 6,090
               units of lot Q1729, 18,1690 units of lot Q1733, 14,210 units
               of lot Q1737.
REASON         The Mini-Set was manufactured with a 12" Macrobore upper
               tubing section instead of a Microbore upper tubing section
               as labeled which also increases the priming volume by
               approximately 1 ml.  The custom set and Uni set have kinking
               tubing and pinholes that may compromise package integrity.

_______________
PRODUCT        Hafcon brand Head Activated Fluoroscopic Controller,
               mechanism for regulating fluoroscopic infared light
               transmitters.  Recall #Z-297-4.
CODE           All serial numbers.
MANUFACTURER   Electrotechnics, Corporation, Shreveport, Louisiana. 
RECALLED BY    Manufacturer, December 7, 1993.  Firm-initiated recall
               ongoing.

                                    -7-DISTRIBUTION   Louisina Indiana, Ohio, Texas, Virginia.
QUANTITY       35 units.
REASON         The firm does not have any manufacturing documents showing
               that the product was made to specification.

_______________
PRODUCT        IVAC Needleless IV Sets, used for the administration of IV
               solutions.  Recall #Z-301-4.
CODE           All lots.
MANUFACTURER   IVAC Corporation, San Diego, California.
RECALLED BY    Manufacturer, by telephone November 18, 1994.  Firm-
               initiated recall complete.
DISTRIBUTION   California, Kansas, Ohio, South Carolina, Utah, Canada,
               England, France, New Zealand.
QUANTITY       13,900 units were distributed.
REASON         Leaking caused by defective manufacturing material.

_______________
PRODUCT        Cobe Centry System 3 Blood Tubing Set, Catalog #003110-500. 
               Recall #Z-302-4.
CODE           Lot numbers:  07Y15317, 07Y15318.
MANUFACTURER   CGH Medical, Inc., Tijuana, Mexico.
RECALLED BY    Cobe Renal Care, Inc., Lakewood, Colorado, by telephone
               November 9, 1993, and November 30 through December 3, 1993,
               followed by letter February 11, 1994.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide, Bermuda, Santiago, Chile, Taiwan.
QUANTITY       21,390 sets were distributed.
REASON         The solvent used to bond the tubing with the luer was
               contaminated with alcohol during manufacturing due to a
               manufacturing operator error which could result in a
               disconnection of the tubing at the luer.

_______________
PRODUCT        Ohmeda Rascal II Anesthetic Gas Monitor, a dedicated
               respiratory gas monitoring system that identifies anesthetic
               agents and measures inspired and expired oxygen as well as
               nitrogen.  Recall #Z-303-4.
CODE           Serial numbers:  FZAU00001 to FZAU00394, FZAV00395 to
               FZAV00886 to FZAW00887 to FZAW01697, with software versions
               1.10 or lower.
MANUFACTURER   Ohmeda Medical Systems Division, Salt Lake City, Utah.
RECALLED BY    Ohmeda Medical Systems Division, Madison, Wisconsin, by
               letters dated January 17 and 18, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide, Canada.
QUANTITY       1,222 units were distributed.
REASON         The anesthetic gas monitoring equipment could misidentify
               the anesthetic gas after long periods of continuous use, and
               if the misidentification involves halothane, the displayed a
               concentrate could be twice the actual concentration, due to
               a defect with software version 1.10 or lower.

                                    -8-_______________
PRODUCT        Baxter Auto Syringe Model AS40A Infusion Pump, product code
               1M8565, Rx medical device used for accurate continuous or
               intermittent infusion delivery of fluids and drugs via
               intravenous, intra-arterial, epidural, or subcutaneous
               routes of administration.  Recall #Z-305-4.
CODE           All units with D1 or D2 decal adjacent to the serial number.
MANUFACTURER   Baxter Healthcare Corporation PTE Ltd., Singapore.
RECALLED BY    Baxter Healthcare Corporation, I.V. Systems Division, Round
               Lake, Illinois, by letters dated February 14, 1994.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,993 units were distributed.
REASON         The infusion pump's optional drug library contains drug
               delivery protocols which may be in conflict with the drug
               manufacturers' package insert protocols for the drugs. 
               Confusion over and misinterpretation of various protocols
               presented in the insert and in the drug library could result
               in over infusion or under infusion of fluids/medications. 
               Also, concentration of fluids could be faulty.

_______________
PRODUCT        Stryker Auxiliary Video Organizer, Catalog #240-085-000.
               Recall #Z-306-4.
CODE           All serial numbers.
MANUFACTURER   Stryker Endoscopy, San Jose, California.
RECALLED BY    Manufacturer, by letter March 25, 1993.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       9 carts.
REASON         The bottom frame of the cart was bent.

_______________
PRODUCT        Gonochek II, used for in-vitro diagnostic for the rapid
               identification of pathogenic Neisseria species.  
               Recall #Z-372-4.
CODE           Product code 13-003-25.
MANUFACTURER   EY Laboratories (HK), Ltd., Tsuen Wan, New Purritories, Hong
               Kong.
RECALLED BY    EY Laboratories, Inc., San Mateo, California, December 17,
               1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       4,845 tests.
REASON         The test will result in a false negative result when used
               for isolates of Neisseria species grown on certain media.

_______________
PRODUCT        Epics Brand Cytometers and Dye Laser:
               (a)  Epics C Series Flow Cytometer; Catalog Nos. 6602370,
               6602808, 6603235, 6602741, 6603376, 6602809, 6602812,
               6602813, 6603365, 6603375, 6603534/6603538, 6603440,
               6603530/6603533, 6603670; 

                                    -9-               (b) Epics 500 Series Flow Cytometer; Catalog Nos.           
               6602837, 66022838, 6603207, 6603102/6603104,
               6603489/6603493; 
               (C) Epics 700 Series Flow Cytometer; Catalog Nos. 6602666,
               6602669, 6602674, 6602121, 6602317, 6602557,
               6603166/6603168, 6603430, 6603669, 6603054, 6603666,
               6603169, 6603171, 6603431, 6603187, 6603246, 6603247,
               6603248, 6603432, 6602645, 6602661, 6603172, 6603173,
               6603433, 6603655; 
               (d) Epics Dye Laser; Catalog No. 6603234; 
               (e) Epics IV Flow Cytometer; Catalog Nos. 6601394, 
               6602295; 
               (f) Epics V Flow Cytometer; Catalog Nos. 6601395, 6601396,
               6602541; 
               (g) Epics Elite Flow Cytometer; Catalog Nos.
               6604042/6604045.  Recall #Z-381/387-4.
CODE           All serial numbers.
MANUFACTURER   Coulter Corporation, Epics Group, Miami Lakes, Florida.
RECALLED BY    Coulter Corporation, Miami Florida, by letter January 21,
               1992.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       Undetermined.
REASON         There is a potential for electrical shock which may result
               into burns to the operator.

_______________
PRODUCT        LC Jet D Disposable Nebulizer used to administer/inhale a
               prescribed dosage of medication.  Recall #Z-461-4.
CODE           Lot #011293.
MANUFACTURER   Paul Ritzau Pari-Werk, GmbH, Starnberg, Germany.
RECALLED BY    Pari Respiratory Equipment, Inc., Richmond, Virginia, by
               telephone January 26 and 27, 1994, followed by letter dated
               January 28, 1994.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       5,302 units were distributed.
REASON         A portion of the dose of the drug is ejected as small
               droplets through the filler port at the top of the device
               rather than being nebulized out the side of the device as
               intended.

_______________
PRODUCT        Model MGU-10A Mammography System, used for mammographic
               studies.  Recall #Z-465-4.
CODE           None.
MANUFACTURER   Toshiba Corporation, Tokyo, Japan.
RECALLED BY    Toshiba America Medical Systems, Inc., Tustin, California.
               FDA approved the firm's corrective action plan March 29,
               1994.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       11 units.

                                   -10-REASON         Device failed to comply with the Federal Performance
               Standard for Diagnostic X-Ray System and their major
               components for user information as required.

_______________
PRODUCT        Stiegmann-Goff Endoscopic Ligator Kits, for endoscopic
               ligation of varices and hemorrhoids:
               (a) Stiegmann-Goff Endoscopic Ligator, Catalog #200220 (5
               bands);
               (b) Stiegmann-Goff Endoscopic Ligator, Catalog #200221 (10
               bands);
               (c) Stiegmann-Goff Endoscopic Ligator, Catalog #200219, 1
               per box.  Recall #Z-529/531-4.
CODE           All lots.
MANUFACTURER   Superior Health Care Group, Inc., Cumberland, Rhode Island.
RECALLED BY    Bard Interventional Products, Tewksbury, Massachusetts, by
               letter December 3, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       37,710 ligators and 3,414 overtubes were distributed.
REASON         No 510(k) has been filed for the Esophageal use of the
               ligator and no 510(k) was filed for the overtube used as an
               accessory to the endoscopic procedure.

_______________
PRODUCT        Allergan Medical Optics (AMO) Prestige Pack Disposable
               Tubing Sets: 
               (a) Model OPO-30S Phaco I/A Tubing Set, Small Bore, provides
               items necessary to perform one irrigation/aspiration
               procedure or one phacoemulsification procedure with AMO
               cataract equipment;
               (b) Model OPO-30C Phaco I/A Tubing Set, provides items
               necessary to perform one irrigation/aspiration procedure or
               one phacoemulsification procedure with AMO cataract
               equipment;
               (c) Model OPO-40, Phaco I/A Tubing Set, provides items
               necessary to perform one irrigation/aspiration procedure or
               one phacoemulsification procedure with AMO prestige advanced
               cataract extraction system.  Recall #Z-532/534-4.
CODE           All lots manufactured since 12/91 as follows:  M8514, M8517,
               M2566, M8679, M9247, M9350, M9365, M9527, M9558, M9559,
               M9700, M9725, M9753, M9915, M10217, M10279, M10375, M11074,
               M11192, M11221.
MANUFACTURER   Allergan Medical Optics, subsidiary of Allergan, Inc.,
               Lenoir City, North Carolina.
RECALLED BY    Allergan Medical Optics, Division of Allerlgan, Inc.,
               Irvine, California, by telephone November 9, 1993, followed
               by letter November 10, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 175,790 sets; (b) 12,036 sets; (c) 390 sets were
               distributed; firm estimates the following amounts remain on
               the market: (a) 7,324 sets; (b) 1,504; (c) 97.

                                   -11-REASON         Cracks have been found on outer trays used to package
               sterile tubing sets.  The cracks result from excessive
               brittleness of the high-impact polystyrene after gamma
               irradiation, therefore compromising the sterility of the
               tubing product inside the package.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        SMA Reference Reagent, intended for in-vitro diagnostic use
               in the calibration of SMA analyzers.  Recall #Z-443-4.
CODE           Lot #181180 EXP 6/94.
MANUFACTURER   Amresco, Inc., Solon, Ohio.
RECALLED BY    Manufacturer, by telephone September 3, 1993, followed by
               letter. Firm-initiated recall complete.
DISTRIBUTION   California, North Dakota, Canada.
QUANTITY       100 vials.
REASON         The device was found to have been assigned an incorrect lot
               number.

_______________
PRODUCT        NAMIC Angiographic Guidewires, intended for use in the
               percutaneous introduction of catheters:
               (a) Catalog #96000105; (b) Catalog #96000106; (c) Catalog
               #96000204; (d) Catalog #96000207; (e) Catalog #96000208; 
               (f) Catalog #96000209.  Recall #Z-444/449-4.
CODE           Lot numbers:  (a) 01-043580; (b) 01-043564; (c) 01-043512,
               02-043512; (d) 01-043568, (e) 01-043569, (f) 01-043555.
MANUFACTURER   Lake Region Manufacturing Company, Inc., Chaska, Minnesota.
RECALLED BY    NAMIC (North American Instrument Company), Glens Falls, New
               York, by telephone January 11, 1993, followed by letter
               January 12, 1994.  Firm-initiated recall ongong.
DISTRIBUTION   Nationwide and international.
QUANTITY       5,575 units were distributed.
REASON         Labeling does not correctly identify the shape of the tip.


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS I
============
_______________
PRODUCT        (a) Phase II Horse Feed; (b) Phase III Horse Feed;
               (c) Horse and Pony Feed.  Recall #V-034/036-4.
CODE           Manufacturing codes are dates of production:  11/9/93,
               11/11/93, 11/23/93, 12/10/93, 1/14/94.
MANUFACTURER   Molinos de Puerto Rico, Inc., Guaynabo, Puerto Rico.
RECALLED BY    Manufacturer, by visit February 7, 1994.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Puerto Rico.
QUANTITY       Horse and Pony Feed - 48,000 pounds; Phase Two and Phase
               Three Horse Feed - 155,000 pounds.
REASON         Possible cross-contamination of horse feed with monensin.

                                   -12-
SEIZURES:
=================================================================
_______________
PRODUCT        Love Land Food Supplement, Good News for Pets Food
               Supplement, Love Land Conditioning Shampoo, Good News for
               Pets Skin and Coat Treatment, accompanying labeling, and the
               drug component precipitated sulfur (93-603-143/6).
CHARGES        Adulterated - The articles are not generally reconized by 
               qualified experts as safe and effective for their intended
               use and there are no approved applications in effect. 
               Misbranded - One of the articles is a drug which is
               fabricated from two or more ingredients and its label fails
               to bear the established name and quantity of each
               ingredient, nor does it bear adequate directions for its
               intended use for the treatment of dermatitis.  All of the
               articles were manufactured in an establishment not duly
               registered and they were not included in a required list.
LOCATION       Normera Resources Corporation, Syracuse, New York.
FILED          February 15, 1994; U.S. District Court for the Northern 
               District of New York; Civil #94-CV-197(FJS); FDC #66903.
SEIZED         March 11, 1994 - goods valued at approximately $22,000.

_______________
PRODUCT        Oxygen USP, Helium USP, Nitrous Oxide USP, and Carbon
               Dioxide USP (94-710-343, et al). 
CHARGE         Adulterated - The methods used in, and the facilities and 
               controls used for, the articles' manufacture and processing,
               do not conform to and are not operated and administered in
               conformity with current good manufacturing practice
               regulation.  
LOCATION       BOC Airco Gases, Carol Stream, Illinois.
FILED          March 29, 1994; U.S. District Court for the Northern
               District of Illinois, Eastern Division; Civil #94C 1948;
               FDC #66937.
SEIZED         March 29, 1994 - goods valued at approximately $195,000.

_______________
PRODUCT        Love Land Food Supplement, Good News for Pets Food
               Supplement, Love Land Conditioning Shampoo, Good News for
               Pets Skin and Coat Treatment, accompanying labeling, and the
               drug component precipitated sulfur (93-603-143/6).
CHARGES        Adulterated - The articles are not generally reconized by 
               qualified experts as safe and effective for their intended
               use and there are no approved applications in effect. 
               Misbranded - One of the articles is a drug which is
               fabricated from two or more ingredients and its label fails
               to bear the established name and quantity of each
               ingredient, nor does it bear adequate directions for its
               intended use for the treatment of dermatitis.  All of the
               articles were manufactured in an establishment not duly
               registered and they were not included in a required list.
LOCATION       Normera Resources Corporation, Syracuse, New York.

                                   -13-FILED          February 15, 1994; U.S. District Court for the Northern 
               District of New York; Civil #94-CV-197(FJS); FDC #66903.
SEIZED         March 11, 1994 - goods valued at approximately $22,000.

_______________
PRODUCT        Oxygen USP, Helium USP, Nitrous Oxide USP, and Carbon
               Dioxide USP (94-710-343, et al). 
CHARGE         Adulterated - The methods used in, and the facilities and 
               controls used for, the articles' manufacture and processing,
               do not conform to and are not operated and administered in
               conformity with current good manufacturing practice
               regulation.  
LOCATION       BOC Airco Gases, Carol Stream, Illinois.
FILED          March 29, 1994; U.S. District Court for the Northern
               District of Illinois, Eastern Division; Civil #94C 1948;
               FDC #66937.
SEIZED         March 29, 1994 - goods valued at approximately $195,000.

                                   -14-
END OF ENFORCEMENT REPORT FOR APRIL 6, 1994.  BLANK PAGES MAY FOLLOW.
                                   ####