FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
11/10/1993

Recalls and Field Corrections:  Foods -- Class I -- 11/10/1993

November 10, 1993                                             93-45

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I
==========================
_______________
PRODUCT           Fresh ready-to-eat cooked crabmeat in 1 pound plastic
                  containers:  (a) Special Crabmeat; (b) Lump Crabmeat. 
                  Recall #F-051/052-4.
CODE              None.  All crabmeat processed during the period of 8/11/93-
                  9/10/93.
MANUFACTURER      Simonson Seafood Ltd., Warsaw, Virginia.
RECALLED BY       Manufacturer, by telephone August 31, 1993 and September
                  24, 1993.  Firm-initiated recall complete.
DISTRIBUTION      Pennsylvania, Virginia.
QUANTITY          Firm estimates none remains on the market.
REASON            Products are contaminated with Listeria monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
=========================
_______________
PRODUCT           Nectarines, fresh, raw, US Grades No. 1 and No. 2, packed
                  in 20 pound cardboard boxes.  Recall #F-533-3.
CODE              0817 EOO G 32.
MANUFACTURER      Ray Beauleaurer, Beauleaurer Farms, Grandview, Washington
                  (Grower).
RECALLED BY       E.W. Brandt & Sons, Wapato, Washington, by telephone August
                  31, 1993.  Firm-initiated recall complete.
DISTRIBUTION      Minnesota.
QUANTITY          50 boxes.
REASON            Product contained illegal residues of Dimethoate.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
========================
_______________
PRODUCT           Dressings and condiments:
                  (a) Salad Dressing, in 1 gallon plastic jars;
                  (b) Lite Salad Dressing, in 32 ounce glass jars;
                  (c) Tarter Sauce, in 8 ounce glass jars and in 1 gallon
                  plastic jars;
                  (d) Cole Slaw Dressing, in 1 gallon plastic jars;
                  (e) Thousand Island Dressing, in 1 gallon plastic jars;
                  (f) Ollie Sauce, in 1 gallon plastic jars;
                  (g) Burger Tarter Sauce, in 8 ounce glass jars.
                  Recall #F-006/012-4.
CODE              Lot numbers:  (a) S3522, S3532, S3562, S3572, S3642, S0043,
                  S0053, S0063, S0083, S0263, S0323;
                  (b) S0153; (c) S0043; (d) S0063; (e) S3532; (f) S3582,
                  S0213; (g) S0063.
MANUFACTURER      Silver Foods, Corporation, Louisville, Kentucky.
RECALLED BY       Manufacturer, by telephone on or around January 1993 -
                  February 1993.  Firm-initiated recall complete.
DISTRIBUTION      Illinois, Ohio, Kentucky, Virginia, California, Wisconsin,
                  Tennessee.
QUANTITY          3,736 cases were distributed; firm estimates none remains
                  on the market.
REASON            Products were contaminated with mold and yeast.
 
_______________
PRODUCT           "Astro for Bison" Lowfat Yogurt and Apple Crisp Granola, in
                  4.4 ounce plastic containers.  Recall #F-013-4.
CODE              All code dates up to and including SEP 1 93.
MANUFACTURER      Bison Foods Company, Buffalo, New York.
RECALLED BY       Manufacturer, by letter dated July 26, 1993, followed by
                  telephone.  Firm-initiated recall complete.
DISTRIBUTION      New York, Pennsylvania, New Jersey, Maryland, Vermont.
QUANTITY          1,808 cases were distributed; firm estimates none remains
                  on the market.
REASON            Product contains undeclared sulfites.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT           Atenolol with Chlorthalidone Tablets, 50 mg/25 mg, Rx used
                  as a combination antihypertensive/diuretic, packaged in
                  bottles of 100.  Recall #D-018-4.
CODE              Lot numbers:  25322 EXP 7/95 (Rugby & Mutual), 25323 EXP
                  7/95 (Mutual, URL & Geneva), 25324 EXP 7/95 (URL, Goldline,
                  Mutual).
MANUFACTURER      Mutual Pharmaceutical Company, Inc., Philadelphia,
                  Pennsylvania.
                                      -2-RECALLED BY       Manufacturer, by letter September 15, 1993.  Firm-initiated
                  recall complete.
DISTRIBUTION      Nationwide.
QUANTITY          Lot #25322 -- 1,662 bottles (Rugby), 1,675 bottles
                  (Mutual); Lot #25323 -- 352 bottles (Mutual), 144 bottles
                  (URL), 4,320 bottles (Geneva); Lot #25324 -- 288 bottles
                  (URL), 49 bottles (Goldline), 809 bottles (Mutual).
REASON            Product does not meet disintegration specifications.

_______________
PRODUCT           Compressed Medical Oxygen, USP, in size E cylinders.
                  Recall #D-019-4.
CODE              12 cylinders without lot numbers.  11 with the following
                  serial numbers and one cylinder with an unknown serial
                  number:
                  Serial Number       Hydrostatic Testing Date 
                  0191A021852                12/16/92 
                  0191A0221819               12/16/92  
                  16694                      12/16/92  
                  0890A017230                12/16/92  
                  0191A021903                12/16/92  
                  15642                      2/01/93  
                  6293A060712                3/22/93  
                  15015D                     7/15/93  
                  0011042                    8/09/93  
                  0293A060706                8/25/93 
                  IT5647                     9/03/93.  
MANUFACTURER      County Health Services, Inc., doing business as American
                  Medical Equipment (AME), Georgetown, Texas.
RECALLED BY       Manufacturer, by visit between September 22-28, 1993. 
                  Firm-initiated recall ongoing.
DISTRIBUTION      Texas.
QUANTITY          Firm estimates none remains on the market. 
REASON            Current good manufacturer practice deficiencies.

_______________
PRODUCT           Lacrisert, a sterile Rx 5 mg Hydroxypropyl Cellulose
                  Ophthalmic Insert and Applicator, indicated in patients
                  with moderate to severe dry eye syndromes.  Recall #D-020-
                  4.
CODE              Lot #W0213.
MANUFACTURER      Merck Manufacturing Division, Division of Merck & Company,
                  Inc., West Point, Pennsylvania.
RECALLED BY       Manufacturer, by letters beginning November 4, 1993.  Firm-
                  initiated recall ongoing.
DISTRIBUTION      Nationwide and international.
QUANTITY          4,077 units were distributed; firm estimates that very
                  little if any of the product remains on the market.
REASON            Lack of assurance of sterility.

                                      -3-_______________
PRODUCT           Cortisone Acetate Tablets USP, 10 mg and 25 mg, in bottles
                  of 100, used as replacement therapy in adrenocortical
                  deficiency states.   Recall #D-021/022-4.
CODE              Lot numbers:  982DF, 733HS (10 mg); 561FD, 616HJ, 535DW
                  (25 mg).
MANUFACTURER      The Upjohn Company, Kalamazoo, Michigan.
RECALLED BY       Manufacturer, by letters December 4, 1992 and November 8,
                  1993.  Firm-initiated recall ongoing.
DISTRIBUTION      Nationwide and international.
QUANTITY          29,790 bottles of lot numbers 982DF, 733HS, 561FD and 616HJ
                  and 14,539 bottles of lot numbers 535DW were distributed.
REASON            Product does not meet dissolution specifications through
                  expiration date.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT           Olympus PK 7100 Data Diskettes, used to store data from the
                  hard drive of the Olympus PK 7100 Automated Pretransfusion
                  Blood Testing System.  Recall #B-032-4.
CODE              Catalog #ZM0472; Lot P111.
MANUFACTURER      Formats Unlimited, East Farmingdale, New York.
RECALLED BY       Olympus America, Inc., Lake Success, New York, by letter
                  June 2, 1993.  Firm-initiated recall complete.
DISTRIBUTION      Nationwide.
QUANTITY          46 diskettes.
REASON            Diskettes were defective.

_______________
PRODUCT           (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
                  Recall #B-034/036-4.
CODE              Unit #04K34211.
MANUFACTURER      American Red Cross Blood Services, Dedham, Massachusetts.
RECALLED BY       Manufacturer, by letter September 24, 1992.  Firm-initiated
                  recall complete.
DISTRIBUTION      Massachusetts.
QUANTITY          1 unit of each component.
REASON            Blood products, which tested negative for the antibody to
                  human immunodeficiency virus type 1 (anti-HIV-1), but were
                  collected from a donor who previously tested repeatedly
                  reactive for anti-HIV-1, were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT           Software programs:  (a) UPS Version 10.22 DOS; (b) Unix
                  Systems, used for the re-entering of catheter description
                  points in the Nucletron Planning System.
                  Recall #Z-046/047-4.
CODE              All Nucletron Planning Systems.
MANUFACTURER      Nucletron B.V., Waardegelder 1, The Netherlands.

                                      -4-RECALLED BY       Nucletron Corporation, Columbia, Maryland, by sending
                  Technical Notes to all NPS users on October 15, 1992. 
                  Firm-initiated recall ongoing.
DISTRIBUTION      Nationwide, Brazil, Canada, West Indies.
QUANTITY          231 units.
REASON            In a situation with implants of two or more cathetrs, an
                  error situation can occur when the catheters are described
                  with catheter description points.  

_______________
PRODUCT           Adams Compact II Centrifuge, distributed as a Primary Care
                  Diagnostic product and through several private label
                  distributors and clinical laboratories which include:
                  (a) Catalog #0225 - Becton Dickinson
                  (b) Catalog #0226 - Allied Clinical Laboratories
                  (c) Catalog #0230 - Methpath
                  (d) Catalog #0231 - Roche
                  (e) Catalog #0232 - Damon Clinical Laboratories
                  (f) Catalog #0234 - Mawd Medical Laboratories
                  (g) Catalog #0236 - Eastside Medical Laboratories
                  (h) Catalog #0237 - Laboratory of Clinical Medicine
                  (i) Catalog #0238 - Metwest Clinical Laboratories
                  (j) Catalog #0239 - Clinical Diagnostic Services.
                  Recall #Z-064/073-4.
CODE              Serial numbers:  25101 through 262348.  
MANUFACTURER      Becton Dickinson, Sparks, Maryland.
RECALLED BY       Manufacturer, by letter March 24, 1993.  Firm-initiated
                  recall ongoing.
DISTRIBUTION      Nationwide.
QUANTITY          Approximately 13,177 units were distributed; firm estimates
                  2,215 units remain on the market.
REASON            Under certain operating conditions, the rotating tube
                  shields may contact the wiring within the housing resulting
                  in abraded insulation or exposed wires.

_______________
UPDATE            Dilipan Hygroscopic Cervical Dilator, listed in the
                  November 3, 1993 Enforcement Report should read Recall #Z-
                  035-4.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT           Dermatology laser systems:
                  (a) Candela SPTL-1a Dermatology Laser System (P/N
                  9904-10-0300/0302); Candela SPTL-1aG (Germany) Dermatology
                  Laser System (P/N 9904-10-0302);
                  (b)  Candela SPTL IEC Dermatology Laser System (P/N
                  9904-0010-0250).  Recall #Z-018/019-4.
CODE              (a) SPTL-1a (9904-10-0300), Serial Nos. 2106, 2219, 
               2220, 3109-3149, 3151-3164, 3167, & 3169 
               SPTL-1aG (Germany) (9904-10-0302), Serial Nos. 
               3150, 3165, 3166, 3168, 3170, 3171, 3173, 3174;

                                      -5-                  (b) SPTL IEC (9904-10-0250), Serial Nos. 2217, 
               2244-2267, 2269-2290.
MANUFACTURER      Candela Laser Corporation, Wayland, Massachusetts.
RECALLED BY       Manufacturer, by letter March 17, 1993.  Firm-initiated
                  field correction ongoing.
DISTRIBUTION      Nationwide and international.
QUANTITY          (a) 64 units; (b) 54 units.
REASON            The design allowed for possible internal fires.

_______________
PRODUCT           Techmate 1000, an automated immunostaining system for
                  tissue samples placed on microscope slides.
                  Recall #Z-061-4.
CODE              All serial numbers are prefixed with alphabetic code TMR
                  followed by 10001, 10006-13, 10015-25, 10027-33, 10035-43.
MANUFACTURER      Biotek Instruments Inc., Santa Barbara, California.
RECALLED BY       Manufacturer, by letter dated June 24, 1993.  Firm-
                  initiated recall ongoing.
DISTRIBUTION      California, Colorado, Connecticut, Illinois, Massachusetts,
                  Maryland, Michigan, Minnesota, North Carolina, New Jersey,
                  New York, Ohio, Oregon, Pennsylvania, Texas, Washington,
                  Vermont, District of Columbia.
QUANTITY          40 units.
REASON            The device's mechanical arm will stop after the failure of
                  an electronic component which senses the position of the
                  mechanical arm.  The component failure is caused by an
                  improperly designed resistor.

_______________
PRODUCT           Bioscan Reagent Strips Dip and Read Test for measuring
                  glucose, ketone, bilirubin, protein, pH and blood in urine,
                  in bottles each holding 100 test strips.  Recall #Z-062-4.
CODE              Lot 28202.
MANUFACTURER      Yeong Dong Pharmaceutical Corporation, Seoul, Korea.
RECALLED BY       Dia-Screen Corporation, Minneapolis, Minnesota, by letter
                  September 20, 1993.  Firm-initiated recall ongoing.
DISTRIBUTION      Massachusetts, Michigan, Ohio.
QUANTITY          120 bottles.
REASON            The test strips are not detecting blood urine in the
                  specified time.  The test strips are intended to detect 5-
                  15 blood cells per microliter one minute after being dipped
                  in a positive urine sample.

_______________
PRODUCT           Ortho brand, Round Bottom, MRC-5 Human Embryonic Lung
                  (Diploid) Cell Culture tubes, used for the isolation of
                  infectious agents and as an indicator system for the
                  presence of toxins.  Recall #Z-063-4.
CODE              Product #505050, lot #063050 EXP 7/29/93.
MANUFACTURER      Ortho Diagnostic Systems, Inc., Carpinteria, California.
RECALLED BY       Manufacturer, by telephone on June 28 & 29, 1993.  Firm-
                  initiated recall complete.

                                      -6-DISTRIBUTION      California, Connecticut, Illinois, Maine, Missouri, New
                  Jersey, New York, Ohio, Oklahoma, Oregon.
QUANTITY          1,995 tubes were distributed; firm estimates none remains
                  on the market.
REASON            Tubes are contaminated with the organism Xanthonas
                  (pseudomonas) Maltophilia which causes the device not to
                  function as designed. 


MEDICAL DEVICE SAFETY ALERTS: 
============================================
_______________
PRODUCT            Acufex Orthopedic Screwdrivers:
                   (a) Part #013108, 7 mm Cannulated Driver, Large Handle;
                   (b) Part #013109, 9 mm Cannulated Driver, Large Handle;
                   (c) Part #013426, 7 mm Cannulated Driver, Small Handle;
                   (d) Part #013427, 9 mm Cannulated Driver, Small Handle.
                   Safety Alert #N-063/064-3 & N-087/088-3.
CODE               Lot numbers:  (a) LCN 591, LCN 691, LCN 791, LCN 891, LCN 991, LCN
                   1091, LCN 1191, LCN 1291, LCN 1391, LCN 192, LCN 292, LCN 392, LCN
                   492, LCN 592, LCN 692, LCN 792, LCN 892, LCN 992, LCN 1092, LCN 1192,
                   LCN 1292, LCN 1392, LCN 1492.
                   (b) LCN 391, LCN 491, LCN 591, LCN 691, LCN 791, LCN 891, LCN 192, LCN
                   292, LCN 392, LCN 492, LCN 592, LCN 692, LCN 792, LCN 892; (c) LCN
                   491, LCN 591, LCN 691, LCN 791, LCN 891, LCN 991, LCN 1091, LCN 192,
                   LCN 292, LCN 392, LCN 492, LCN 592, LCN 692, LCN 792, LCN 892, LCN
                   992, LCN 1092, LCN 1192, LCN 1292, LCN 1392, LCN 1492, LCN 1592, LCN
                   1692, LCN 1792; (d) LCN 491, LCN 591, LCN 691, LCN 791, LCN 192, LCN
                   292, LCN 392, LCN 492, LCN 592, LCN 692, LCN 792, LCN 892, LCN 992,
                   LCN 1092, LCN 1192.
MANUFACTURER       Acufex Microsurgical, Inc., Mansfield, Massachusetts.
ALERTED BY         Manufacturer, by letter dated October 15, 1992.
DISTRIBUTION       Nationwide and international.
QUANTITY           (a) 1,235 units; (b) 793 units; (c) 734 units; (d) 454. 
REASON             There is a potential for tip breakage.


SEIZURE:
==================================================================
______________
PRODUCT            Pickled gouramy fish (93-683-331).
CHARGES            Adulterated - The article consists in part of a filthy substance by
                   reason of the presence therein of insects, and rodent, human, and
                   cat/dog hair.
FIRM               Cathay Company, Los Angeles, California.
FILED              October 21, 1993; U.S. District Court for the Central District of
                   California: Civil #93-6380LBG; FDC #66788.
SEIZED             November 2, 1993 - goods valued at approximately $25,000.

                                             -7-
END OF ENFORCEMENT REPORT FOR NOVEMBER 10, 1993.  BLANK PAGES MAY
FOLLOW.
                                            ####