FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
07/02/1992

RECALLS AND FIELD CORRECTIONS: July 2, 1992

                             FOODS - CLASS I
                                                               92-27
              
PRODUCT        Su Turkish Apricots, dried, in 10 pound boxes.  Recall #F-336-2.
CODE           All product distributed prior to 3/12/92.
MANUFACTURER   Blooming Prairie Natural Foods, Minneapolis, Minnesota.
RECALLED BY    Manufacturer, by letter March 27, 1992.  Firm-initiated recall   
               complete.
DISTRIBUTION   Minnesota, Iowa, South Dakota, Wisconsin.
QUANTITY       Approximately 700 pounds were distributed.
REASON         Product contains undeclared sulfites.

              
PRODUCT        Salinas brand Vacuum Packed Smoked Milkfish, 1 to 1.5 pounds.
               Recall #F-337-2.
CODE           None.
MANUFACTURER   Salinas Foods Import, Woodside, New York.
RECALLED BY    Manufacturer, by letter April 9, 1992.  Firm-initiated recall    
               ongoing.
DISTRIBUTION   New York, New Jersey.
QUANTITY       Approximately 200 cases (35 pieces per case) were manufactured;  
               firm estimates 25 cases remain on the market.
REASON         Potential for Clostridium botulinum toxin.
                                       
              
PRODUCT        Captain D's brand Ready-To-Eat Frozen Seafood Salad, in 1-1/2    
               pound aluminum pans.  Recall #F-351-2.
CODE           None.
MANUFACTURER   Shaw's Southern Belle Frozen Foods, Inc., Jacksonville, Florida.
RECALLED BY    Manufacturer, by telephone February 11, 1992.  Firm-initiated    
               recall complete.
DISTRIBUTION   Southeast United States.
QUANTITY       2 cases were distributed.
REASON         Product was contaminated with Listeria monocytogenes.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II
              
PRODUCT        Harvest of Eden brand Mushrooms, pieces and stems, in 68 ounce   
               cans (drained weight).  Recall #F-350-2.
CODE           Three line code:  MPS/TC/1Y28.
MANUFACTURER   Tak Chance, Hong Kong.
RECALLED BY    Rema Foods, Inc., Teaneck, New Jersey, by letter May 13, 1992.   
               Firm-initiated recall ongoing.
DISTRIBUTION   New York, New Jersey, Pennsylvania, Ohio, Oklahoma, Wisconsin,   
               Maine.
QUANTITY       1,600 cases (6 cans per case) were distributed.
REASON         Product is contaminated with Staphylococcus enterotoxin.

              
PRODUCT        Various flavors of cookie dough:
               (a) Original Cookie Dough
               (b) Pecan Cookie Dough
               (c) Oatmeal Cookie Dough
               (d) Double Fudge Cookie Dough
               (e) Brownie Cookie Dough
               (f) Peanut Butter Cookie Dough
               (g) White Macadamia Cookie Dough
               (h) Dark Macadamia Cookie Dough
               (i) Cheesecake Cookie Dough
               (j) Sugar Cookie Dough
               (k) Jumbo Cookie Dough
               (l) Caramel Pecan Cookie Dough.  Recall #F-352/363-2.
CODE           Dates including and beginning with 011392 through 020292.
MANUFACTURER   Great American Chocolate Chip Cookie Company, Inc., Atlanta,     
               Georgia.
RECALLED BY    Manufacturer, by telephone February 19, 1992.  Firm-initiated    
               recall complete.
DISTRIBUTION   Nationwide, Gaum.
QUANTITY       21,946 cases were distributed; firm estimates none remains on    
               the market.
REASON         Product was contaminated with metal particles.

              
PRODUCT        SSIPS brand Awesome Orange Coolers.  Recall #F-364-2.
CODE           Lot 097A1 printed on bottom of container.
MANUFACTURER   Johanna Farms, Inc., Flemington, New Jersey.

                                      -2-
RECALLED BY    Manufacturer, by telephone May 26 & 27, 1992, general news       
               media May 28 & 29, 1992, and by letter May 29, 1992.             
               Firm-initiated recall ongoing.
DISTRIBUTION   New Jersey, Pennsylvania, New Hampshire, Delaware.
QUANTITY       8,640 units were distributed.
REASON         Product is contaminated with sanitizing solution (sodium         
               hydroxide).


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
              
PRODUCT        Various food supplements manufactured and distributed in bulk    
               containers:  (a) High Protein Chocolate Tablets; (b) High        
               Protein Vanilla Tablets; (c) Gain Weight Vanilla Tablets; (d)    
               Special Tablets; (e) One Daily Vitamin Supplement with Iron      
               Tablets.  Recall #F-343/347-2.
CODE           Lot numbers 8762 through 8791 for food supplement products and   
               lot #56203 for the vitamin supplement with iron.
MANUFACTURER   High Chemical Company, Philadelphia, Pennsylvania.
RECALLED BY    Manufacturer, by telephone May 20, 1992.  Firm-initiated recall  
               complete.
DISTRIBUTION   Unknown.
QUANTITY       Firm estimates none remains on the market.
REASON         Products were contaminated with rodent filth.

              
PRODUCT        4C All Natural brand Grated Parmesan Cheese, in 8 ounce          
               paperboard containers.  Recall #F-348-2.
CODE           OCT259341, OCT259342.
MANUFACTURER   4C Foods Corporation, Brooklyn, New York.
RECALLED BY    A.A. Green Company, Inc., also known as Hester & Connley,        
               Jacksonville, Florida, by visit beginning March 2, 1992.         
               Firm-initiated recall complete.
DISTRIBUTION   Florida, Georgia.
QUANTITY       11,232 containers were distributed; firm estimates none remains  
               on the market.
REASON         Product was contaminated with mold.

              
PRODUCT        Nobility brand Mandarin Oranges, 11 ounces.  Recall #F-349-2.
CODE           All product with P on the top of the can.
MANUFACTURER   Unknown.  Product of People's Republic of China.
RECALLED BY    Mitsui Foods, Hackensack, New Jersey (importer), by telephone    
               March 1, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Southeastern states from North Carolina to Florida.
QUANTITY       11,524 cases (24 cans per case) were distributed.
REASON         Product was contained in swollen cans.

                                      -3-
RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS I
              
PRODUCT        Metaproterenol Sulfate Inhalation Solution, USP 5%,              
               Sulfite-Free; a Rx oral inhalation bronchodilator for treatment  
               of bronchial asthma; in 1 fluid ounce (30 ml) vial, for use      
               with a hand-bulb nebulizer or IPPB apparatus, packaged under     
               the following labels:  PBI, Goldline, H.L. Moore, Qualitest,     
               Rugby, Dey.  Recall #D-397-2.
CODE           All unexpired lots (13 lots).
MANUFACTURER   Pharmaceutical Basics, Inc. (PBI), Morton Grove, Illinois.
RECALLED BY    Manufacturer, by letter May 27, 1992 and June 1, 1992.           
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       100,781 units were distributed; firm estimates 20,000 units      
               remain on the market.
REASON         Microbial growth, Pseudomonas gladioli/cepacia.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
              
PRODUCT        Pro-Banthine Tablets, propantheline bromide USP, 7.5 mg tablets  
               and 15 mg tablets, a Rx oral tablet used as adjunctive therapy   
               in the treatment of peptic ulcer, packaged as follows:
               (a) 7.5 mg in 100 tablet bottles;
               (b) 15 mg in 100 tablet unit dose boxes.  Recall #D-399/400-2.
CODE           Lot numbers:  (a) 0990-418 and 0990-419 EXP 9/92;
               (b) 0590-483 and 0590-485 EXP 5/93.
MANUFACTURER   Searle & Company, Inc., Caguas, Puerto Rico
RECALLED BY    Schiapparelli Searle, a joint venture of Searle                  
               Pharmaceuticals, Inc., Skokie, Illinois, by letter dated June    
               19, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       16,510 bottles and 1,383 unit dose cartons were distributed.
REASON         Subpotency.

              
PRODUCT        Various Rx injectable drug products:
               (a) Gentamicin Sulfate Injection, USP (Pediatric), 10 mg/ml, in  
               2 ml multiple dose vials, an injectable antibiotic;
               (b) Heparin Lock Flush Solution, USP, 100 USP units/ml, for      
               cleaning intermittent infusion sets;
               (c) Lidocaine HCl Injection, USP, 1%, 10 mg/ml, in 2 ml single   
               dose vials, a local anesthetic for infiltration and nerve        
               block.  Recall #D-401/403-2.
CODE           Lot numbers:  (a) 320148 EXP 2/94; (b) 311367 EXP 12/93;
               (c) 320057 EXP 7/93.
MANUFACTURER   Lyphomed, Division of Fujisawa USA, Inc., Grand Island,
               New York.

                                      -4-
RECALLED BY    Lyphomed, Division of Fujisawa USA, Inc., Melrose Park,          
               Illinois, by letter June 12, 1992.  Firm-initiated recall        
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 87,775 vials were distributed; firm estimates 10,000 vials   
               remain on the market; (b) 241,100 vials were distributed; firm   
               estimates 60,000 vials remain on the market; (c) 120,900 vials   
               were distributed; firm estimates 61,000 vials remain on the      
               market.
REASON         The vials have a low fill volume.

              
UPDATE         An UPDATE listed in the June 24, 1992 Enforcement Report for     
               Recall #D-365/371-2 should have read:
               Various injectable drug products, manufactured by Lyphomed,      
               Division of Fujisawa USA, Inc., Recall #D-365/371-2, which       
               appeared in the June 10, 1992 Enforcement Report has been        
               extended as follows:  Heparin Sodium Injection, 5,000 USP units  
               has been extended to include lot #320179 EXP 2/94 and Sodium     
               Chloride Injection, USP, 0.9% has been extended to include lot   
               #320140 EXP 1/94.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
              
PRODUCT        Diazoxide Injection, USP, 15 mg/ml, in 20 ml single dose vials,  
               a Rx antihypertensive.  Recall #D-404-2.
CODE           Lot numbers:  120026 EXP 7/93 and 111249 EXP 6/93.
MANUFACTURER   Lyphomed, Division of Fujisawa USA, Inc., Melrose Park,          
               Illinois. 
RECALLED BY    Manufacturer, by letter June 17, 1992.  Firm-initiated recall    
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       4,761 vials were distributed; firm estimates 2,380 vials remain  
               on the market.
REASON         The pH is below USP limits prior to expiration date.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS I
              
PRODUCT        (a) Model 2000W Heart/Respiration Monitors;
               (b) Model 2000W Heart/Respiration Monitors with Option H;
               Model 700 Recorder/Monitors.  Recall #Z-708/710-2.
CODE           Serial numbers:
               (a) 4195     4309     6451     20179     20218
                   20596    20701    20746    20932     21003
                   21004    21300    21359    21387     21390
                   21410    21635    21906    22455     22512
                   22554    22626    22772    22818     23003
                   23472    23528    23616    23661     23950
                   24334    24610    24814    20497     20589
                   21338    21644    21832    22196     22159
                   23181    23182    23548    22901     23778
                   23838    24580    24603

                                      -5-
               (b) 4513     5349     4861     6279      4753
               (c) 70001.
MANUFACTURER   EdenTec, Eden Prairie, Minnesota.
RECALLED BY    Manufacturer by telephone beginning May 1, 1992, followed by     
               letter May 4, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY       53 monitors and 1 recording device containing a monitor were     
               distributed.
REASON         The devices contain a defective circuit board which may cause    
               the apnea alarm to sound inadequately.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
              
PRODUCT        CDI Cells, flow through devices inserted directly into the       
               extracorporeal tubing circuit, sterile and individually          
               packaged for cut-in at the time of use or as part of custom      
               tubing pack:
               (a) Model 6631 Pouch; (b) Model 6641 Pouch; (c) Model 6651 Case;
               (d) 6731 Pouch; (e) Model 6741 Pouch; (f) Model 6751 Case.
               Recall #Z-628/633-2.
CODE           All lot numbers of all models.
MANUFACTURER   CDI, 3M Healthcare, Tustin, California.
RECALLED BY    Manufacturer, by telephone beginning December 18, 1991,          
               followed by letters dated and sent December 18, 1991 through     
               January 15, 1992.  Firm-initiated recall complete.
DISTRIBUTION   Alabama, California, Florida, Illinois, Kentucky, Louisiana,     
               Massachusetts, Missouri, Mississippi, North Carolina, New        
               Jersey, Nevada, New York, Texas, Virginia, West Virginia.
QUANTITY       (a) 1,510 units; (b) 2,460 units; (c) 1,390 units; (d) 1,370     
               units; (e) 1,290 units; (f) 70 units were distributed; firm      
               estimates none remains on the market.
REASON         The film/paper pouch used to package these devices may be        
               punctured or damaged during normal shipping and handling,        
               compromising the sterility of the devices.

              
PRODUCT        SpermPrep medium, a sterile solution used for the in-vitro       
               washing of sperm before it is introduced into the uterus:
               (a) SpermPrep Medium, in 12 ml glass vials, Catalog #SPM-12;
               (b) SpermPrep Medium, in 100 ml glass bottles, Catalog #SPM-100.
               Recall #Z-805/806-2.
CODE           Lot numbers:  (a) BSA-4151; (b) BSA-4152.
MANUFACTURER   Quelab Laboratories, Inc., Montreal, Canada.
RECALLED BY    Fertility Technologies, Natick, Massachusetts, by telephone      
               December 8, 1991 followed by letter December 27, 1991.           
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Spain.
QUANTITY       (a) 251 units; (b) 5 units were distributed.
REASON         The sterility of the product has been compromised as evidenced   
               by loss of package integrity or lack of adequate                 
               sterilization.  Mold was found on the rubber stopper of the 12   
               ml vial.

                                      -6-
             
PRODUCT        Nicolet Viking Electrodiagnostic Systems, used to perform        
               routine and computer-aided electromyography, evoked potential    
               studies, and nerve condition studies:
               (a) Nicolet Viking; (b) Nicolet Viking LE; (c) Nicolet Viking    
               II.  Recall #Z-812/814-2.
CODE           Serial numbers:  (a & b) 860106 through 919188 (not inclusive);
               (c) 900101 through 920748 (not inclusive).
MANUFACTURER   Nicolet Instrument Corporation, Madison, Wisconsin.
RECALLED BY    Manufacturer, by letter dated April 10, 1992.  Firm-initiated    
               field correction ongoing.
DISTRIBUTION   Nationwide, Belgium, Canada, France, Germany, Japan, Mexico,     
               The Netherlands, Switzerland, The United Kingdom.
QUANTITY       Approximately 1,066 units were distributed.  All units shipped   
               after February 6, 1992 have been modified.
REASON         Under certain conditions, it is possible that patient burns can  
               occur at electrode sites.

              
PRODUCT        MRT-35 Magnetic Resonance Imaging System using Software Version  
               4.01C, used to facilitate the interface of the MRT-35 operator   
               and the system.  Recall #Z-815-2.
CODE           Serial numbers range from Q-0007 to Q00276.
MANUFACTURER   Toshiba America MRI, Inc., South San Francisco, California.
RECALLED BY    Manufacturer, by letter March 30, 1992.  Firm-initiated field    
               correction complete.
DISTRIBUTION   Nationwide, Germany, Switzerland, People's Republic of China.
QUANTITY       142 software systems were distributed.
REASON         Software Version 4.01C contains a defect whereby an image that   
               is displayed on the screen prints the correct run number and     
               image, but not the correct patient name.

              
PRODUCT        X-Ray Controls for Fluoroscopic X-Ray Systems with Pulsed        
               Fluoroscopy Adaptor PFA-01A:
               (a) Model KXO-2050; (b) Model KXO-80 A/B;
               (c) Model KXO-80-C/D.  Recall #Z-819/821-2.
CODE           All serial numbers.
MANUFACTURER   Toshiba Medical Systems, Inc., Tustin, California.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action plan    
               June 10, 1992.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       150 units were distributed.
REASON         The recording device was not interlocked with the higher         
               exposure rates as required.

              
PRODUCT        7200 Series Microprocessor Ventilator, used to enable a          
               respiratory care practitioner to provide a patient with          
               ventilatory assistance.  Recall #Z-822-2.
CODE           All 7000 series models, (7200AE, 7200SPE, 7200E) and all serial  
               numbers of ventilators shipped from 9/9/91 to 1/31/92.

                                      -7-
MANUFACTURER   Puritan-Bennett Corporation, Carlsbad, California.
RECALLED BY    Manufacturer, by letter March 20, 1992.  Firm-initiated recall   
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,309 units were distributed.
REASON         A 5.0 amp fuse instead of a 0.5 amp fuse was installed in the    
               ventilator's utility panel harness, which could potentially      
               cause a diode to open, resulting in loss of power alarm and      
               overheating of electrical components.

              
PRODUCT        Cryo High Pressure Hoses, intended for those users that do not   
               have the Cryo carrier (use with CR4000 consoles), used for       
               connecting the CR4000 consoles to either a nitrous oxide or      
               carbon dioxide non-syphon gas cylinder:
               (a) 60" Cryo High Pressure Hose, Part #P350003;
               (b) 30" Cryo High Pressure Hose, Part #P350007.
               Recall #Z-829/830-2.
CODE           Lot #139366.
MANUFACTURER   Furon Company, Synflex Division, Mantua, Ohio.
RECALLED BY    Mira, Inc., Waltham, Massachusetts, by telephone February 14,    
               1992 followed by letter dated April 3, 1992.  Firm-initiated     
               recall ongoing.
DISTRIBUTION   Nationwide, Brazil, Australia, South Africa, Peru, Canada,       
               Syria, Egypt, Bulgaria, Japan, Austria.
QUANTITY       (a) 55 units; (b) 15 units were distributed
REASON         The outer jacket of the hose was not properly perforated to      
               allow the gas to leave the hose.

              
PRODUCT        (a) de-Stat 3 Liquid Cleaning/Disinfecting and storage           
               solution, in 4 fluid ounce and 1 fluid ounce plastic bottles,    
               for fluoro/silicone acrylate and silicone acrylate (REP)         
               contact lenses;
               (b) Sof/Pro-Clean (s.a.) daily cleaner for soft (hydrophilic)    
               contact lenses, in 1 fluid ounce plastic bottles.
               Recall #Z-831/832-2.
CODE           Lot numbers and EXP dates:
               (a) 0870 EXP 4/93, 1020 EXP 4/93, 1100 EXP 5/93, 1840 EXP 7/93,  
               2070 EXP 7/93, 2370 EXP 8/93, 2410 EXP 8/93, 2810 EXP 10/93,     
               3020 EXP 10/93, 3520 EXP 12/93, 0621 EXP 3/94, 0921 EXP 4/94,    
               0991 EXP 4/94, 1831 EXP 7/94, 2121 EXP 8/94, 2331 EXP 8/94;
               (b) 0022 EXP 1/95.
MANUFACTURER   Sherman Pharmaceuticals, Inc., Abita Springs, Louisiana.
RECALLED BY    Manufacturer, (a) by letter October 15, 1991; (b) by letter      
               April 16, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Israel, Korea.
QUANTITY       Approximately (a) 115,000 units; (b) 6,815 units were            
               distributed.
REASON         The de-Stat product was mislabeled.  The firm also removed the   
               sodium bisulfite component from the Sof/Pro-Clean product        
               without an approved supplement to its PMA.


                                      -8-
              
UPDATE         Contact lenses under the following names:
               (a) NewVues; (b) Focus; (c) Cibasoft Softcolors and Visitint;
               (d) Cibathin.  Recall #Z-801/804-2, which appeared in the June   
               24, 1992 should read:
RECALLED BY    Manufacturer, by letter June 5, 1991.  Firm-initiated recall     
               completed November 1991 with return product destroyed.
QUANTITY       Firm estimates none remains the market.
REASON         NewVues are mislabeled as having a base curve of 8.4 when the    
               actual curve was 8.8, Focus lenses indicate the wrong power on   
               the carton, and some broken vial pieces were found in vials of   
               Cibasoft Softcolors, Visitint and Cibathins.


MEDICAL DEVICE SAFETY ALERT
              
PRODUCT        (a) Multi-sized Double Lumen Percutaneous Intra-Aortic Balloons  
               (IABs):  30 cc (8F), 40 cc (9F), 50 cc (10F), 40 cc (10.5F);
               (b) Double Lumen IAB for Sheatless:  40 cc (9F S);
               (c) Flexi-Cath Lumen Catheter:  9.5F.
               Safety Alert #M-057/059-2.
CODE           All lot numbers before 3/12/91.
MANUFACTURER   Kontron Instruments, Inc., Everett, Massachusetts.
ALERTED BY     Manufacturer, by letter May 22, 1992.
DISTRIBUTION   Nationwide and international.
QUANTITY       17,407 units were distributed; firm estimated that 2,611 units   
               remained on the market as of 9/30/91.
REASON         The firm's May 22, 1992 letter informs users of a problem with   
               helium leaks at the bifurcation/metal luer of the device and     
               provides the user with information including problem             
               recognition and correction.


                                      -9-

END OF ENFORCEMENT REPORT FOR JULY 1, 1992.  BLANK PAGES MAY FOLLOW.
                                     ###