FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
05/20/1992

RECALLS AND FIELD CORRECTIONS: May 20, 1992

                                FOODS -- CLASS I

                                                                           92-21

              
PRODUCT        Eight patterns of porcelain dinnerware, labeled in Chinese       
               characters and marked MADE IN CHINA:
               (a) Blue flower pattern consisting of plates, spoons, bowls,     
               soup plates, and condiment dishes;
               (b) Red Long-Life Traditional Pattern, consisting of plates,     
               bowls, soup plates, and condiment dishes;
               (c) Red Long-Life Modern Pattern, consisting of plates and       
               saucers;
               (d) Red and White Flower Pattern, consisting of plates and       
               condiment dishes;
               (e) Phoenix Pattern, consisting of spoons, plates, saucers, and  
               condiment bowls;
               (f) Two Fish (also known as Goldfish) Pattern, consisting of     
               condiment bowls, plates, and spoons;
               (g) Lovebird Pattern, consisting of spoons, plates, and bowls;
               (h) Old Man and Deer (also known as Longevity) Pattern,          
               consisting of plates.  Recall #F-288/295-2.
                                       
CODE           All eight patterns are labeled "MADE IN CHINA."  In addition,    
               the Lovebirds and Red Long Life Traditional Patterns have one    
               of the following three digit numbers stamped on the bottom of    
               the dinnerware:  Lovebirds:  "006/024/068/078/150/184/232/276;
               Red Long Life Traditional:  025/028/080/090/184/194/208/228/
               428/480.  Cardboard cartons are coded as follows:
               (a) 415; (b) 75229; (c) Not coded; (d) 6308; (e) 6326;
               (f) 683114; (g) 681184; (h) 5268.
MANUFACTURER   Product of China.
RECALLED BY    May Wah Trading Company, Inc., San Francisco, California         
               (distributor/dealer, by letter December 9, 1991.                 
               Firm-initiated recall ongoing.
DISTRIBUTION   California, Washington state, Nevada, Colorado.
QUANTITY       Unknown.
REASON         Product contains excess levels of leachable lead.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II

              
PRODUCT        Various flavors of sherbet ice cream:  (a) Lemon Sherbet,        
               packaged in 1 pint plastic containers under the Nancy Grey       
               label, packaged in 3 gallon cardboard containers, no brand name;
               (b) Lime Sherbet, packaged in 1 pint plastic containers under    
               the Nancy grey label, packaged in 3 gallon cardboard             
               containers, no brand name;
               (c) Rainbow Sherbet, packaged in 1/2 gallon paper containers     
               under Reinhold's Richtest label, packaged in 1-1/2 gallon        
               plastic containers, no brand, packaged in 3 gallon cardboard     
               containers, no brand name.  Recall #F-297/299-2.
CODE           Not coded.  All stock of all three products.
MANUFACTURER   Reinhold Ice Cream Company, Pittsburgh, Pennsylvania.
RECALLED BY    Manufacturer, by letter April 29, 1992.  Firm-initiated recall   
               ongoing.
DISTRIBUTION   Pennsylvania.
QUANTITY       Unknown.
REASON         Products contain undeclared FD&C Yellow No. 5.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III
              
PRODUCT        Betty Ann Creecy Greens (dry land cress), in 16 ounce cans.      
               Recall #F-296-2.
CODE           Lot numbers:  CH1M1/CRES, CH1M2/CRES, CL1M1/CRES, CL1M2/CRES,    
               CM1M1/CRES, CM1M2/CRES, CN1M1/CRES, CR1M1/CRES, CR1M2/CRES,      
               CS1M1/CRES, CS1M2/CRES, CT1M1/CRES, CT1M2/CRES, CV1M1/CRES.
MANUFACTURER   Monticello Canning Company, Inc., Crossville, Tennessee.
RECALLED BY    Manufacturer, by telephone October 30, 1991, followed by letter  
               October 31, 1991.  Firm-initiated recall complete.
DISTRIBUTION   Maryland, Tennessee, Virginia, Kentucky, North Carolina, West    
               Virginia.
QUANTITY       11,433 cases (24 cans per case) were distributed.
REASON         Product is contained in cans with seam defects.

                                      -2-
RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II

              
PRODUCT        Minoxidil Tablets, USP, (a) 2.5 mg; (b) 10 mg, in bottles of     
               100 and 500, an Rx antihypertensive.  Recall #D-339/340-2.
CODE           All sizes and All codes.
MANUFACTURER   Royce Labs., Inc., Miami, Florida.
RECALLED BY    Manufacturer, by letter April 22, 1992.  Firm-initiated recall   
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 22,730 100-tablet bottles; (b) 55,384 100-tablet bottles     
               and 6,469 500-tablet bottles were distributed.
REASON         Lack of assurance of bioequivalence and abbreviated new drug     
               application discrepancies.

             
PRODUCT        (a) Bausch & Lomb brand Pilocarpine HCl 2% Solution, in 15 ml    
               bottles, sterile Rx ophthalmic;
               (b) Hydroproplymethyl Cellulose and Dextran 70 Solution Sterile  
               Lubricant Eye Drops, in 15 ml bottles.  Recall #D-343/344-2.
CODE           Lot numbers:  (a) 0312; (b) 0363.
MANUFACTURER   Pharmafair, Inc., Hauppauge, New York.
RECALLED BY    Manufacturer, by letter October 24, 1991.  Firm-initiated        
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       (a) 8,886 bottles; (b) 3,030 bottles were distributed; firm      
               estimates none remains on the market.
REASON         Lack of assurance of sterility.

              
PRODUCT        Ephedrine Hydrochloride 50 mg Sustained Release Tablets, in 20   
               tablet and 60 tablet packets, promoted for weight loss.
               Recall #D-345-2.
CODE           Not coded.  All product.
MANUFACTURER   G & J Enterprise, Pascagoula, Mississippi (repacker).
RECALLED BY    Repacker, by visit between April 20 and 24, 1992.                
               Firm-initiated recall ongoing.
DISTRIBUTION   Mississippi.
QUANTITY       Firm estimates 2,000 tablets remain on the market.
REASON         Product marketed without new drug approval.

              
PRODUCT        Various Rx and OTC repacked drugs, in unit-dose packages.        
               Recall #D-346-2.
CODE           Products within expiration period which had been unit-dose       
               repacked by Pharmaceutical Services of Middle Tennessee, Inc.
MANUFACTURER   Pharmaceutical Services of Middle Tennessee, Inc., Nashville,    
               Tennessee (repacker).
RECALLED BY    Repacker, by telephone March 24, 1992, followed by letter dated  
               March 25, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Tennessee.
QUANTITY       721,869 repacked units were distributed.

                                      -3-
REASON         Current good manufacturing practice deficiencies.

              
PRODUCT        Azamacort Inhaler, 20 gram, metered-dose aerosol unit, an Rx     
               drug for the control of the symptoms of bronchial asthma.        
               Recall #D-348-2.
CODE           Lot #BHP060.
MANUFACTURER   Sterling Pharmaceutical, Barceloneta, Puerto Rico.
RECALLED BY    Rhone Poulenc Rorer Pharmaceuticals, Inc., Fort Washington,      
               Pennsylvania, by telephone April 28, 1992, followed by letter    
               April 29, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       31,552 units were distributed.
REASON         Presence of foreign oil matter.

              
PRODUCT        Fluphenazine Decanoate Injection, 25 mg/ml, in 5 ml multiple     
               dose vials, an Rx SVP antipsychotic.  Recall #D-350-2.
CODE           Lot numbers:  110727 EXP 5/92, 110740 EXP 5/92, 110745 EXP       
               5/92, 110787 EXP 6/92, 110807 EXP 6/92, 110835 EXP 6/92, 110880  
               EXP 6/92, 110910 EXP 7/92.
MANUFACTURER   Lyphomed, Division of Fujisawa USA, Inc., Melrose Park,          
               Illinois. 
RECALLED BY    Manufacturer, by letter May 8, 1992.  Firm-initiated recall      
               ongoing.
DISTRIBUTION   Nationwide and Puerto Rico.
QUANTITY       27,152 units were distributed.
REASON         Subpotency.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
              
PRODUCT        Oxygen USP, Compressed Medical Gas transfilled into E, D, 40,    
               50 and 60 cu foot cylinders.  Recall #D-341-2.
CODE           None.
MANUFACTURER   Olympic Ambulance, Inc., doing business as Olympic Oxygen        
               Service, Sequim, Washington.
RECALLED BY    Manufacturer, by telephone beginning April 30, 1992.             
               Firm-initiated recall ongoing.
DISTRIBUTION   Washington state.
QUANTITY       20 cylinders were distributed.
REASON         Current good manufacturing practice deficiencies.

              
PRODUCT        Acetaminophen with Codeine Tablets, 300 mg acetaminophen and 60  
               mg codeine phosphate per tablet, a schedule C-III narcotic oral  
               analgesic, unit dose packaged with 20 tablets per card.  Recall  
               #D-342-2.
CODE           Lot #2C704 EXP 3/94.

                                      -4-
MANUFACTURER   Barr Laboratories, Inc., Pomona, New York.
RECALLED BY    UDL Laboratories, Inc., Rockford, Illinois (repacker), by        
               telephone April 7, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   California, New Jersey, Michigan, Pennsylvania.
QUANTITY       111 unit cartons of 100 tablets were distributed; firm           
               estimates none remains on the market.
REASON         Product fails content uniformity specifications.

              
PRODUCT        Maximum Strength Enteric Coated Aspirin Tablets, 500 mg,         
               packaged in 60 tablet bottles, under the following labels:       
               Hook's, Peoples, Brooks, United Supermarket, Fays, Schein,       
               Taylor, Rite Aid, Thrifty, Silver Saver.  Some bulk product was  
               also involved which was sold to Great Lakes Co., but shipped to  
               Crown Drug Co. for repacking.  Recall #D-347-2.
CODE           Lot numbers:  K6019E, J6074D, J6073D (Hook's);
               K6107D, K6019E, J6074D, K6220D, H6237D, J6073D, J6072D, A6040E   
               (Peoples); H6112D, H6237D, H6111D (Brooks); H6112D (United       
               Supermarket); K6019E, J6074D, J6073D (Fays); K6019E, J6074D,     
               K6220D, H6235D (Schein); H6111D, H6112D, H6113D (Taylor);        
               H6112D, H6113D, H6235D, H6236D, H6237D, J6072D (Rite Aid);       
               A6041E, K6019E, J6074D, K6220D, H6235D, J6073D, H6111D, 6041E    
               (Thrifty); 6041E (Silver Saver); K6106D, K6107D, K6108D,         
               K6109D, K6219D, K6220D, L6189D, L6190D, L6191D, A6040E (bulk).
MANUFACTURER   LNK International, Hauppauge, Long Island, New York.
RECALLED BY    Pennex Products Company, Inc., Verona, Pennsylvania, by          
               telephone and by letter dated March 18, 1992.  Firm-initiated    
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 30,000 bottles were distributed.
REASON         Product fails dissolution specifications.

              
PRODUCT        Amiloride HCl and Hydrochlorothiazide 5 mg/50 mg, in bottles of  
               100 and 1,000, used for hypertension, under the following        
               labels:  Royce, Goldline, Genetco, Martec, Qualitest Products,   
               and United Research Labs.  Recall #D-349-2.
CODE           Lot numbers IJ1127, IJ1128, IJ1129, IJ1130.
MANUFACTURER   Royce Labs, Inc., Miami, Florida.
RECALLED BY    Manufacturer, by letter April 23, 1992.  Firm-initiated recall   
               ongoing.
DISTRIBUTION   Florida, New York, Missouri, Alabama, Pennsylvania.
QUANTITY       3,288 1,000-tablet bottles and 31,888 100-tablet bottles were    
               distributed under the private labels.  Amount distributed under  
               own (Royce) label is unknown.
REASON         Use of unapproved processor for hydrochlorothiazide raw          
               material. 

              
PRODUCT        OTC Analgesic Combination marketed as "HI-VALUE Relief for Pain  
               adult strength pain reliever analgesic", in 100 tablet           
               bottles.  Recall #D-351-2.
CODE           Lot 901037.

                                      -5-
MANUFACTURER   C-B Laboratories, Inc., Harrisburg, North Carolina.
RECALLED BY    Manufacturer, by telephone.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Texas.
QUANTITY       6,264 bottles were distributed.
REASON         Product degradation.

              
UPDATE         Recall #D-321-2, which appeared in the April 29, 1992            
               Enforcement Report should read:  Amphotericin B for Injection,   
               USP, Preservative Free, 50 mg/vial.  


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
              
PRODUCT        Source Plasma.  Recall #B-289-2.
CODE           Unit numbers:  E-46047-091, E-46339-091, E-46389-091,            
               E-46415-091, E-46453-091, E-46490-091, E-46531-091, E-46568-091.
MANUFACTURER   Community Plasma of Dallas, Inc., Casa Grande, Arizona.
RECALLED BY    Manufacturer, by telephone and facsimile August 5, 1991.         
               Firm-initiated recall ongoing.
DISTRIBUTION   California.
QUANTITY       8 units.
REASON         Source Plasma, which tested negative for hepatitis B surface     
               antigen (HBsAg) and the antibody to human immunodeficiency       
               virus type 1 (anti-HIV-1), but were collected from a donor who   
               previously tested reactive for either HBsAg or anti-HIV-1, were  
               distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
              
UPDATES        Recall #B-192-2, Reagent Red Blood Cells Resolve Panel B Cell    
               #22, which appeared in the April 8, 1992 Enforcement Report is   
               a Class III Recall.

               Recall #B-219-2, Surgiscreen Reagent Red Blood Cells, which      
               appeared in the April 8, 1992 Enforcement Report should read:
               MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New      
               Jersey.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
              
PRODUCT        Lifestream Centrifugal Pump Motor Drive, Model 2100 MD, used     
               during cardiopulmonary bypass procedures.  Recall #Z-259-2.
CODE           LOt #501230-03.
MANUFACTURER   St. Jude Medical, Inc., Chemlsford, Massachusetts.
RECALLED BY    Manufacturer, by telephone.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida, Texas, China.
QUANTITY       3 units were distributed.

                                      -6-
REASON         The device may fail due to failure of the motor drive.           
               Reworked units may have had magnet pole failure which caused     
               the motor rotor to lock-up.

              
PRODUCT        Laparoscopy Graspers, hand held manual instruments used during   
               laparoscopic surgery:
               (a) LA1300, Large Jaw Grasper, Monopolar, serrated large jaw     
               for grasping or blunt dissection;
               (b) LA1305, Alligator Jaw Grasper with Ratchet, insulated        
               shaft, used for gallbladder retraction;
               (c) LA1310, Small Jaw Dissector, Monopolar, rounded serrated     
               jaw for dissection and grasping;
               (d) LA1320, Bullet Nose Dissector, Monopolar, conical, serrated  
               jaw for fine dissection in difficult areas;
               (e) LA1340, Long Jaw Dissector, Monopolar, long, tapered,        
               serrated jaws for precise dissection in a variety of sites;
               (f) LA1350, Atraumatic Grasper/ Dissector, Monopolar, smooth,    
               rounded jaws for atraumatic grasping or dissection.
               Recall #Z-465/470-2.
CODE           Lot numbers QG, QH, QI, QJ for catalog LA1305; All lots, QA      
               thru RC for catlog numbers LA1300, LA1310, LA1320, LA1340,       
               LA1350.
MANUFACTURER   Baxter Healthcare Corporation, St. Louis, Missouri.
RECALLED BY    Baxter Healthcare Corporation, V. Mueller Division, Niles,       
               Illinois, by letters dated February 20, 1991 and March 30,       
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Canada, Finland, Italy, Belgium.
QUANTITY       (a) 96; (b) 859; (c) 470; (d) 253; (e) 466; (f) 227 units were   
               distributed.
REASON         The upper portion of the jaw may break due to weakness of the    
               metal.

              
PRODUCT        Sunquest and Alisun Series Tanning Beds and Booths.  Recall      
               #Z-471/472-2.
CODE           Serial numbers:  None.
MANUFACTURER   Alisun America, Ltd., McHenry, Illinois.
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action plan    
               April 21, 1992.  Firm-initiated field correction ongoing.
DISTRIBUTION   Nationwide
QUANTITY       19,707 units were distributed.
REASON         Noncompliance with the performance standard for sunlamp          
               products in that the label and user instructions contained       
               multiple (10-20) ultraviolet lamp type designations; and failed  
               to have adequate exposure schedules, labels, and user            
               instructions for the lamp types being used in the devices.

              
PRODUCT        Model G150 Laser Rangefinder, used for distance measuring  of    
               various scientific and industrial applications.
               Recall #Z-473-2.

                                      -7-
CODE           Serial numbers:  None.
MANUFACTURER   Optech systems Corporation, Downsview, Ontario, Canada (T.       
               Hasek Associates, Ltd., Reno, Nevada).
RECALLED BY    Manufacturer.  FDA approved the firm's corrective action plan    
               April 20, 1992.  Firm-initiated field correction ongoing.
DISTRIBUTION   California, Massachusetts, Michigan.
QUANTITY       4 units were distributed.
REASON         Noncompliance with the performance standard for laser products   
               in that its emission levels were in excess of the intended       
               class of the product (Class (I).

              
PRODUCT        Simpson Coronary Atherocath SCA-LP, low profile, 7F, a           
               directional coronary atherectomy catheter used for the           
               selective excision and retrieval of atherosclerotic lesions in   
               coronary arteries and vein drafts.  Recall #Z-483-2.
CODE           Catalog #LP-70-40-09.  Lot numbers:  112881, 112891, 112901,
               EXP 4/92.
MANUFACTURER   Devices for Vascular Intervention, Inc., Temecula, California.
RECALLED BY    Devices for Vascular Intervention, Inc., Redwood City,           
               California, by letter January 17, 1992.  Firm-initiated recall   
               ongoing.
DISTRIBUTION   Pennsylvania, Alabama, Florida, Massachusetts, North Carolina,   
               Ohio, California, Illinois, Indiana, New Jersey, Maryland, New   
               Mexico, South Carolina, Washington state, Mississippi,           
               Michigan, Washington, D.C., Germany, The Netherlands.
QUANTITY       135 units were distributed.
REASON         Urethane cracking may result in detachment of the nosecone as    
               well as fragmentation of urethane.

               
PRODUCT        Puritan Bennett Microprocessor Ventilators containing T-Level    
               Software:  (a) Model 7200a and (b) Model 7200sp.
               Recall #Z-673/674-2.
CODE           All units shipped between 11/15/90 and 2/6/91.
MANUFACTURER   Puritan Bennett Corporation, Carlsbad, California.
RECALLED BY    Manufacturer, by letter February 21, 1991.  Firm-initiated       
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       1,295 units were distributed.
REASON         There is an inconsistency between the operator manual labeling   
               and the alarm function.  When the T-level software is in use,    
               the low exhaled tidal volume alarm monitors mandatory breaths    
               in the SIMV mode.

              
PRODUCT        A.O.R. (Alvarado Orthopedic Research) Universal Alignment Arm,   
               Catalog #00-5061-066-00, for use at the beginning of knee        
               replacement operations to align the head (ball) of the femur     
               with the ankle to determine the exact location for placement of  
               the knee components.  Recall #Z-676-2.
CODE           Lot #79887500.

                                      -8-
MANUFACTURER   Carr Metal Products, Indianapolis, Indiana.
RECALLED BY    Zimmer, Inc., Warsaw, Indiana (packager), by letter February     
               19, 1991.  Firm-initiated recall complete.
DISTRIBUTION   California, Wisconsin, West Virginia, Illinois.
QUANTITY       6 units were distributed.
REASON         The graduations of the metal scale used in the alignment arm     
               are not deep enough to show on x-ray.

              
PRODUCT        Adfuse Infusion Pumps, designed to release solutions of drug at  
               the rate of 60 ml per 30, 50, or 90 minutes, depending on the    
               model:  Product #AF6030; (b) Product #AF6050, 7FS001, AF6050-SK;
               (c) Product #6090.  Recall #Z-677/679-2.
CODE           All lot numbers.
MANUFACTURER   Healthtek, Inc., Nevada City, California.
RECALLED BY    Manufacturer, by FAX and letter March 28, 1992.  Firm-initiated  
               recall ongoing.
DISTRIBUTION   Nationwide, Canada, Germany Australia, The Netherlands.
QUANTITY       Approximately 30,000 units were distributed; firm estimates      
               9,500 units remain on the market.
REASON         The devices are subject to leakage, erratic infusion rates, and  
               have been distributed without an approved 510(k) for the use of  
               the drugs labeled or for the design changes.

               
PRODUCT        Puritan Bennett Display Panel, Monitor Screen, an accessory for  
               the Model 7200 Series Microprocessor Ventilator.
               Recall #Z-681-2.
CODE           All Puritan Bennett 7202 Display monitors shipped between        
               October 17, 1990 and March 13, 1991.
DISTRIBUTION   Nationwide and international.
QUANTITY       527 units were distributed; firm estimates 150 units remain on   
               the market.
REASON         Product may contain a defective printed circuit board which may  
               cause the screen to go blank.

              
PRODUCT        Laserex Transcutaneous Electrical Nerve Stimulator (TENS-1000),  
               an Rx device used externally for pain management. 
               Recall #Z-682-2.
CODE           All units.
MANUFACTURER   Laserex Technologies, Unley, Australia.
RECALLED BY    Laserex, Inc., Scottsdale, Arizona, by letter January 31,        
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Arizona, California, Florida, New Jersey, Texas, Mexico.
QUANTITY       1,570 units were distributed.
REASON         The device has been marketed without an approved 510(k)          
               premarket notification.

               
PRODUCT        Medi-Trace ECG electrodes for use as patient monitoring          
               electrodes during electrocardiographic monitoring procedures:    
               (a) Offset Dx ECG Electrodes; (b) S'Offset ECG Electrodes.       
               Recall #Z-683/684-2.
                                      -9-
CODE           All lot numbers between 12181 and 12272.  Specific lot numbers   
               are as follows:  (a)  12181, 12182, 12191, 12192, 12201, 12202,  
               12211, 12212, 12241, 12242, 12251, 12252, 12261, 12262, 12271,   
               12272;
               (b) 12182, 12183, 12191, 12192, 12201, 12202, 12211, 12212,      
               12221, 12241, 12242, 12251, 12252, 12261, 12262, 12271, 12272.
MANUFACTURER   Graphic Controls Corporation, Methuen, Massachusetts.
RECALLED BY    Graphic Controls Corporation, Buffalo, New York, by letter       
               January 7, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       (a) 402,296 electrodes; (b) 321,475 electrodes were distributed.
REASON         The trace quality of the electrodes may be compromised.  The     
               electrodes may produce more noise, artifact or baseline          
               wandering in their tracings than is usual with these products.

              
PRODUCT        Richard Wolf Irrigation and Suction Spatula, Catalog             
               #E8383.801, a medical device designed for the instillation and   
               aspiration of fluid for lavage purposes during laparoscopy, and  
               assisting in the dissection and removal of the gallbladder.      
               Recall #Z-685-2.
CODE           All units.
MANUFACTURER   Richard Wolf Medical Instruments, Vernon Hills, Illinois.
RECALLED BY    Manufacturer, by letter dated January 10, 1992.  Firm-initiated  
               recall ongoing.
DISTRIBUITON   Nationwide.
QUANTITY       107 units were distributed.
REASON         An electrical current may travel to the trumpet valve of the     
               spatula, creating the potential for an electrical shock to the   
               user.

              
UPDATE         Recall #Z-522/523-2, Coulter Automated Hematology Analyzers,     
               Models JT2 and JT3, which appeared in the April 15, 1992         
               Enforcement Report should read:

               REASON:  Due to software anomalies, if the power is              
               interrupted, two conditions may occur upon restoration of power  
               which can cause the device to generate erroneous results.  One   
               result is the dilution of a specimen if the needle has pierced   
               the specimen, but has not retracted when the power is            
               interrupted.  The second is that calibration factors may be      
               lost and the instrument automatically reset to default           
               conditions.  Firm-initiated field correction complete.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
              
PRODUCT        Calibration Gas Cylinder labeled as 7% C02, 7% 02, Balance N2,   
               used in the calibration of blood gas analyzers during            
               production at CIBA-Corning.  Recall #Z-552-2.

                                     -10-
CODE           Lot #150619 EXP 3/92.
MANUFACTURER   Scott Medical Products, Inc., Scott Specialty Gases Division,    
               Plumsteadville, Pennsylvania, by letter February 19, 1992.       
               Firm-initiated recall ongoing.
DISTRIBUTION   Massachusetts.
QUANTITY       1 unit.
REASON         Product actually contains 5% C02, 12% 02, and balance N2.

              
PRODUCT        LEEP RediKit for LOOP Electrosurgical Excision Procedure, a      
               single use prescription device which is used in conjunction      
               with other devices in the removal of lesions of the cervix and   
               similar minor gynecological surgery procedures.
               Recall #Z-675-2.
CODE           Lot numbers:  0691, 10709, 10724, 10805L, 10922L, 11112L,        
               10918L, 10919L, 10920L, 10921L, 10923L, 10924L, 10925L, 11108L,  
               11112L, 11118L, 11119L, 11121L, 11125L, 11127L, 11203L, 11204L,  
               11209L, 11211L, 11214L, 11215L, 20106L, 20115L, 20123L, 20204L,  
               20205L, 20303L, 20306L, 20311L, 20316L.  CooperSurgical label    
               identifies the product as reorder #6060; the Euro-Med private    
               label identifies the product as reorder #LEEP-5.
MANUFACTURER   Astra Pharmaceutical Products, Inc., Westborough,                
               Massachusetts, (supplier of the defective xylocaine (2%) with    
               Epinephrine (1:100,000) component).
RECALLED BY    Cooper Surgical, Inc., Shelton, Connecticut, by letter April     
               3-6, 1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       13,352 kits were distributed.
REASON         A component of the kits, 2% Xylocaine Injection Cartridge with   
               Epinephrine 1:100,000, may contain subpotent epinephrine.  The   
               drug component has been recalled by the manufacturer, ASTRA      
               Pharmaceutical Products, Inc. (Recall #D-286/287-2).


RECALLS AND FIELD CORRECTIONS:  VETERINARY -- CLASS II
              
PRODUCT        Counterfeit Drug Product - 1% Ivomec and Ivomec-F Injection for  
               cattle.  Recall #V-039-2.
CODE           NBC 110 OCT 92.
MANUFACTURER   Tampa Independent Dairy Farmer's Association, Tampa, Florida     
               (distributor).
RECALLED BY    Distributor, by visit in January 1992.  Firm-initiated recall    
               ongoing.
DISTRIBUTION   Florida, Georgia.
QUANTITY       500 bottles were distributed.
REASON         Counterfeit drug product.

              
PRODUCT        Counterfeit Drug Product - 1% Ivomec and Ivomec-F Injection for  
               cattle, in 500 ml bottles.  Recall #V-040-2.
CODE           All codes that contain one or more of the counterfeit            
               characteristics.

                                     -11-
MANUFACTURER   Poultry Health Service of Florida, Inc., Jacksonville, Florida   
               (distributor).
RECALLED BY    Distributor, by visit and by letter during February/March        
               1992.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida.
QUANTITY       1,651 bottles were distributed.
REASON         Counterfeit drug product.


SEIZURES:
              
PRODUCT        Oxygen USP (92-609-471).
CHARGE         Adulterated - The methods used in, and the facilities and        
               controls used for, the manufacture, processing, packing, and     
               holding do not conform to and are not operated and administered  
               in conformity with current good manufacturing practice           
               regulations.
FIRM           Superior Medical Supply, Inc., Villa Park, Illinois.
FILED          May 6, 1992 - U.S. District Court for the Northern District of   
               Illinois, Eastern Division; Civil #92C2980; FDC #66428.
SEIZED         May 7, 1992 - oxygen, valued at approximately $18,764.


DISPOSITIONS:  INJUNCTION
              
AGAINST        Vital Health Products, Ltd., a corporation, and Conrad E.        
               LeBeau, an individual.
PRODUCT        Hydrogen peroxide products (90-564-146, et al.).
CHARGE         New drug - Manufacturing, packaging, or distributing hydrogen    
               peroxide products, and other related products with no approved   
               new drug application in effect for such products;
               Misbranded - Promoting, labeling, advertising, or representing   
               that hydrogen peroxide products are safe and effective in the    
               cure, mitigation, treatment, or prevention of human disease,     
               with no approved new drug application in effect for such
               products.
FILED          April 12, 1991 - Complaint; March 16, 1992 - Decision and        
               Order; U.S. District Court for the Eastern District of           
               Wisconsin; Civ. #91-C-363; INJ #1243.
DISPOSITION    The government's motion for summary judgment was granted and     
               the corporation and its president have been permanently          
               enjoined from promoting and selling hydrogen peroxide products   
               for use in the treatment of serious diseases, including AIDS     
               and cancer.

                                     -12-


END OF ENFORCEMENT FOR MAY 20, 1992.  BLANK PAGES MAY FOLLOW.

                                     ###