ENFORCE
02/26/1992
Recalls and Field Corrections: February 26, 1992
FOODS
Class I - A situation in which there is a reasonable
probability that the use of, or exposure to, a violative
product will cause serious adverse health consequences or
death.
Product: (a) Frozen Razor Clams in 5 pound cans; (b) Minced Razor
Clams, in 7 ounce cans. Recall #F-116/117-2.
Code: (a) All lots produced since 9/1/91; (b) Lot #01118/BMRCL.
Manufacturer: Bell Buoy Crab Company, Seaside, Oregon.
Recalled by: Manufacturer, by telephone November 21-22, 1991.
Firm-initiated recall complete.
Distribution: Washington state, Oregon, California, Virginia, Idaho.
Quantity: (a) 591 cans; (b) 96 cans were distributed.
Reason: Product is contaminated with Domoic Acid.
----
-1-
Class II - A situation in which the use of, or exposure to
a violative product may cause temporary or medically
reversible adverse health consequences or where the
probability of serious adverse health consequences is
remote.
Product: Mehaffies brand (a) Pineapple Pie; (b) Lemon Pies,
32 ounces. Recall #F-120/121-2.
Code: Pies coded with a pull date that expires 4 days from
the date of manufacture.
Manufacturer: Mehaffies Pie Company, Dayton, Ohio.
Recalled by: Manufacturer, by visit July 11, 1991. Firm-initiated
recall complete.
Distribution: Ohio.
Quantity: Approximately (a) 80 pies; (b) 300 pies were distributed;
none remains on market.
Reason: Product contained undeclared FD&C Yellow No. 5.
----
Product: Birds Eye Brussels Sprouts, in 10 ounce foil wrapped
boxes. Recall #F-122-2.
Code: Lot numbers NI0348 5, NI0348 6, NI0348 7.
Manufacturer: Norcal Crosetti Foods, Inc., Watsonville, California.
Recalled by: Kraft General Foods, Inc., Glenview, Illinois, by letter
January 23, 1992. Firm-initiated recall complete.
Distribution: Nationwide.
Quantity: 1,515 cases (24 boxes per case) were distributed; firm
estimates 30 percent remains on the market.
Reason: Product is contaminated with diesel fuel.
----
Product: Kountry Fresh brand (a) Shelled Peanuts; (b) Roasted
Peanuts, in 16 ounce glass jars. Recall #F-123/124-2.
Code: Sell by 5-5-93B.
Manufacturer: John B. Sanfilippo & Son, Inc., Bainbridge, Georgia.
Recalled by: Manufacturer, by visit on or about January 31, 1992.
Firm-initiated recall ongoing.
Distribution: Alabama.
Quantity: Unknown.
Reason: Product contains high aflatoxin levels.
----
Product: Fresh Asparagus imported from Peru, in unlabeled 11
pound cartons. Recall #F-125-2.
Code: None.
Manufacturer: Frutech International Pompnao Beach, Florida.
Recalled by: Manufacturer, by telephone November 20, 1991, followed
by letter November 21, 1991. Firm-initiated recall
complete.
-2-
Distribution: Florida, Michigan, New York, Georgia.
Quantity: 392 cartons were distributed; firm estimates none
remains remain on the market.
Reason: Product is contaminated with Chlorpyrifos.
----
Product: Blended Margarine and Butter Products:
(a) TW Services, 1 pound containers;
(b) Gold-N-Sweet, 5 pound tubs;
(c) Old Fashion, 20 pound pails;
(d) One + Two (1+2), 1 pound containers, and 5 and
50 pound pails;
(e) Fancy Blend, 30 and 50 pound cubes.
Recall #F-127/131-2.
Code: All lot codes that include the letters "CW".
Manufacturer: Wilsey Foods, Inc., City of Industry, California.
Recalled by: Manufacturer, by telephone November 12, 1991 and by
press release issued by the State of California,
Department of Health Services on November 14-15, 1991.
Firm-initiated recall ongoing.
Distribution: Arizona, California, Hawaii, Nevada, Utah.
Quantity: Approximately 1,660,000 pounds were manufactured;
FDA estimates that little product remains in commerce.
Reason: Product is contaminated with Staphylococcus enterotoxin.
----
Class III - A situation in which the use of, or exposure
to a violative product is not likely to cause adverse
health consequences.
CORRECTION: Bottled Water, Recall #F-112-2, manufactured by
Cascade Clear Water Company, Sedro Wooley, Washington,
which appeared in the January 29, 1992 Enforcement
Report is a Class III recall.
----
COSMETICS
Class I -
NONE
Class II -
Product: Hand Lotion, non-medicated, in 12 fluid ounce plastic
pump bottles. Recall #F-126-2.
Code: Lot numbers: MG0390, MG1190, MG1290, MG1390, MH0190
MH0290, MH0990, MH1090, MH1790, MH2090, MH2190.
Manufacturer: Midik Corporation, Elkhart, Indiana.
-3-
Recalled by: Bristol-Myers Squibb Company, Evansville, Indiana, by
voice mail message on October 25, 1991, followed by
sales department bulletin dated October 30, 1991, and
by letter October 31, 1991. Firm-initiated recall ongoing.
Distribution: Nationwide.
Quantity: 84,660 bottles were distributed; firm estimates none
remains on the market.
Reason: Product is contaminated with Pseudomonas Aeruginosa.
----
Class III -
NONE
HUMAN DRUGS AND BIOLOGICS
Class I -
NONE
Class II -
Product: LyphoMed brand Nitroglycerin Injection, USP, 5 mg/ml, in
5 ml and 10 ml vials, a Rx injectable for intravenous
infusion after dilution for control of blood pressure
during surgical procedures. Recall #D-214-2.
Code: Lot numbers: 110359 EXP 10/92 (5 ml), 110367 EXP 10/92
(10 ml).
Manufacturer: Lyphomed, Division of Fujisawa USA, Inc., Melrose Park,
Illinois.
Recalled by: Manufacturer, by letter February 7, 1992. Firm-initiated
recall ongoing.
Distribution: Florida, New York, California, New Jersey, Puerto Rico,
India.
Quantity: 24,250 5-ml vials and 15,720 10-ml vials were distributed;
firm estimates none remains on the market.
Reason: Subpotency.
----
Product: Levothroid (Levothyroxin Sodium, USP) Tablets, 100 mcg,
in 100 tablet bottles, a Rx product used as a replacement
for substitution therapy for diminished or absent thyroid
function. Recall #D-216-2.
Code: 05428-1 EXP 4/93.
Manufacturer: Pharmacy Dispensing Quantities, Inc., Fort Scott, Kansas.
(repacker/responsible firm).
Recalled by: Repacker, by letter December 23, 1991. Firm-initiated
recall ongoing.
Distribution: Nationwide.
Quantity: 1,498 bottles were distributed.
Reason: An Inderal 10 mg tablet was found in a bottle of the
product.
----
-4-
Product: OTC Acetaminophen Liquid, cherry and grape flavors, in
4 fluid ounce plastic bottles, with accompanying dosage
cup, indicated for relief of children's fever and pain,
under the K.C. Pharmaceuticals, Leiner Nutritional
Products, and Topco Associates labels. Recall #D-219-2.
Code: Lot numbers 9B21401, 9C21402, 9F21403, 9F21404, 0A025,
0E048, OK018, 0M016, 0N029, 1D033, 1D035, EXP dates range
from 3/91 to 11/92.
Manufacturer: K.C. Pharmaceuticals, Inc., Pomona, California.
Recalled by: Manufacturer, by letter January 10, 1992. Firm-initiated
recall ongoing.
Distribution: California, Illinois.
Quantity: 81,822 bottles were distributed; FDA estimates 8,000
bottles remain on the market.
Reason: Markings on dosage cup not totally compatible with label
dosage instructions.
----
Product: Platelets Pheresis. Recall #B-121-2.
Code: Unit numbers: 21P27268, 21P27270, 21P27273S, 21P27274S,
21P27274T.
Manufacturer: American Red Cross, Portland, Oregon.
Recalled by: Manufacturer, by telephone June 6, 1991, followed by
letter June 21, 1991. Firm-initiated recall complete.
Distribution: Oregon.
Quantity: 5 units.
Reason: Blood products, labeled with incorrect expiration dates,
were distributed.
----
Product: Recovered Plasma. Recall #B-128-2.
Code: Unit #53R77805.
Manufacturer: American Red Cross Blood Services, Baltimore, Maryland.
Recalled by: Manufacturer, by letter June 26, 1991. Firm-initiated
recall complete.
Distribution: Switzerland.
Quantity: 1 unit.
Reason: Blood product drawn from an unsuitable donor was
distributed.
----
Product: Red Blood Cells. Recall #B-130-2.
Code: Unit #1390990.
Manufacturer: Mississippi Blood Services, Inc., Jackson, Mississippi.
Recalled by: Manufacturer, by telephone October 17, 1991, followed
by letter November 14, 1991. Firm-initiated recall
complete.
Distribution: Mississippi.
-5-
Quantity: 1 unit.
Reason: A unit of red blood cells, testing repeatably reactive
for the antibody to hepatitis C virus encoded antigen
(anti-HCV), was distributed.
----
Product: Red Blood Cells. Recall #B-131-2.
Code: Unit #1678135.
Manufacturer: Walson Army Hospital, Ft. Dix, New Jersey.
Recalled by: Manufacturer, by telephone September 26 or 27, 1991.
Firm-initiated recall complete.
Distribution: New York.
Quantity: 1 unit.
Reason: Blood component which tested repeatably reactive for
hepatitis B surface antigen (HBsAg) was distributed for
transfusion.
----
Product: (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen
Plasma; (d) Recovered Plasma. Recall #B-135/138-2.
Code: Unit numbers: 21-54448, 21-58709, 21-62534, 21-64150,
21-66371, 21-68787, 21-69882, 21-75061, 21-85046, 21-87455,
21-90189, 35-17567, 35-20290, 35-22115, 35-25034, 35-37562,
35-39564, 35-45190, 35-57740, 35-58126, 35-69974, 35-72033,
35-77937, 35-78390, 35-79838, 35-87387, 35-94708, 35-99103,
37-01509, 37-01906, 37-05957, 37-06798, 37-13446, 37-24155,
37-29406, 37-31288;
(b) 35-17567, 35-20290, 35-22115, 35-39564, 35-57740,
35-58126, 35-69974, 35-72033, 35-79838, 35-78390, 35-87387,
35-94708, 35-99103, 37-01509, 37-01906, 37-05957, 37-06798,
37-13446, 37-24155, 37-29406, 37-31288;
(c) 21-85046, 35-20290, 35-37562, 35-76782, 37-01906;
(d) 21-54448, 21-58709, 21-62534, 21-64150, 21-66371,
21-68787, 21-69882, 21-75061, 21-87455, 21-90189, 35-17567,
35-22115, 35-25034, 35-39564, 35-45190, 35-57740, 35-58126,
35-69974, 35-72033, 35-77937, 35-78390, 35-79838, 35-87387,
35-94708, 35-99103, 37-01509, 37-05957, 37-06798, 37-13446,
37-24155, 37-29406, 37-31288.
Manufacturer: Michigan Community Blood Center, doing business as
Grand Valley Blood Program, Grand Rapids, Michigan.
Recalled by: Manufacturer, by telephone November 1, 1990, and by
letters of January 2, 7 and 14, 1991. Firm-initiated
recall ongoing.
Distribution: Mississippi, Virginia, New York, California, New Jersey,
Pennsylvania, Michigan.
Quantity: (a) 36 units; (b) 21 units; (c) 5 units; (d) 32 units.
Reason: Blood components, which tested non-reactive for the
antibody to the human immunodeficiency virus type 1
(anti-HIV-1) but were collected from donors who
previously tested repeatably reactive for anti-HIV-1,
were distributed.
----
-6-
Product: Source Plasma. Recall #B-145-2.
Code: Unit #XF9880.
Manufacturer: Indianapolis Blood Plasma, Inc., Indianapolis, Indiana.
Recalled by: Consignee notified firm October 8, 1991. Recall complete.
Distribution: North Carolina.
Quantity: 1 unit.
Reason: Blood product untested for the antibody to the human
immunodeficiency virus type 1 (anti-HIV-1) and hepatitis B
surface antigen (HBsAg) was distributed.
----
Product: Platelets, Pooled. Recall #B-147-2.
Code: Pool #FF63382.
Manufacturer: The Blood Center of Southeastern Wisconsin, Milwaukee,
Wisconsin.
Recalled by: Manufacturer, by telephone July 25, 1991. Firm-initiated
recall complete.
Distribution: Wisconsin.
Quantity: 1 unit.
Reason: Pooled Platelets, which contained platelets collected
from an ineligible donor due to a reported history of
hepatitis b exposure occurring two weeks prior to
donation, were distributed.
----
Class III -
Product: Rx Tylenol with Codeine No. 3 Tablets, in 100 tablet
bottles. Recall #D-215-2.
Code: Lot #HC1632P EXP 3/94.
Manufacturer: McNeil Pharmaceutical Company, Dorado, Puerto Rico.
Recalled by: McNeil Pharmaceutical, Spring House, Pennsylvania,
by letter January 20, 1992. Firm-initiated recall ongoing.
Distribution: New York, Pennsylvania, Florida, Virginia, Maryland,
Georgia, Alabama, West Virginia, South Carolina, Rhode
Island, Connecticut, North Carolina, Maine, Vermont.
Quantity: 13,680 bottles were distributed; firm estimates 1,000
bottles remain on the market.
Reason: Some bottles in correctly labeled unit cartons were
placed in a packer labeled Tylenol with Codeine No. 2
tablets. The outer shipping carton is correctly labeled.
----
Product: Trandate (Labetalol Hydrochloride) Tablets, 100 mg, in
100 tablet bottles, a Rx drug for use in the management
of hypertension. Recall #D-217-2.
Code: Z60300MX EXP MAY94.
Manufacturer: Pharmacy Dispensing Quantities, Inc., Ft. Scott, Kansas
(repacker/responsible firm).
-7-
Recalled by: Repacker, by letter August 13, 1991. Firm-initiated
recall complete.
Distribution: New York, Missouri, Tennessee, South Dakota, Idaho,
Ohio, Pennsylvania, Indiana.
Quantity: 140 bottles were distributed.
Reason: The label correctly declares the product as 100 mg
but the ingredient statement shows the product as 200 mg.
----
Product: Clonidine HCl Tablets, 0.3 mg, in 100 tablet bottles, a
Rx drug indicated for the treatment of hypertension,
under the Best label only. Recall #D-218-2.
Code: Lot numbers 322991 EXP 8/93, 127491 EXP 9/93.
Manufacturer: Par Pharmaceutical, Inc., Spring Valley, New York.
Recalled by: Manufacturer. Firm-initiated recall ongoing.
Distribution: Florida.
Quantity: 72 bottles were distributed.
Reason: Correct tablet number 112 was misprinted as 110 on the
side panel of the Best brand label.
----
Product: (a) Platelets Pheresis; (b) Source Leukocytes.
Recall #B-119/120-2.
Code: Unit numbers: (a) 21P27407, 21P27409; (b) 21F12431.
Manufacturer: American Red Cross, Pacific Northwest Services,
Portland, Oregon.
Recalled by: Manufacturer, by letter June 24, 1991. Firm-initiated
recall complete.
Distribution: Oregon, Washington state.
Quantity: (a) 2 units; (b) 1 unit.
Reason: Blood products, incorrectly tested for the antibody
to the hepatitis C virus (anti-HCV), were distributed.
----
Product: Red Blood Cells. Recall #B-129-2.
Code: Unit numbers: 5161923 through 5161930, 5161932, 5161933,
5161935, 5161936, 5161939, 5161944, 5161946, 5161949,
5161950, 5161951.
Manufacturer: Department of the Navy, Naval Hospital Camp Lejeune,
Camp Lejeune, North Carolina.
Recalled by: Manufacturer, by telephone August 28-29, 1991. Firm-
initiated recall complete.
Distribution: North Carolina, New Jersey.
Quantity: 18 units.
Reason: Red Blood Cells labeled with incorrect expiration
dates were distributed.
----
-8-
Product: (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma;
(d) Fresh Frozen Plasma; (e) Cryoprecipitated AHF.
Recall #B-139/143-2.
Code: Unit numbers: (a) 21-65779, 21-75172, 32-67143, 35-34999,
35-39374, 35-39472, 35-40820, 35-42151, 35-45886, 35-47282,
35-50970, 35-51753, 35-52477, 35-53952, 35-53986, 35-55003,
35-56697, 35-59369, 35-60827, 35-64630, 35-65208, 35-71771,
35-73412, 35-76782, 35-76793, 35-77105, 35-77285, 35-77492,
35-82167, 35-87288, 35-89389, 35-89831, 35-91751, 35-94573,
35-95222, 35-96137, 35-97931, 35-01399, 37-02147, 37-05105,
37-10576, 37-12827, 37-16040, 37-16873, 37-20114, 37-21703,
37-22351, 37-26799, 37-27245, 37-30921, 37-33717, 37-36371,
37-36838, 37-37409, 41-09170, 41-10558, 41-11004, 41-11898,
41-11946, 41-12982, 41-12989, 41-15625;
(b) 32-67143, 35-34999, 35-39374, 35-39472, 35-40820,
35-42151, 35-45886, 35-47282, 35-50970, 35-51753, 35-52477,
35-53952, 35-53986, 35-55003, 35-56697, 35-59369, 35-60827,
35-64630, 35-65208, 35-71771, 35-73412, 35-76782, 35-76793,
35-77105, 35-77285, 35-82167, 35-87288, 35-89389, 35-91751,
35-94573, 35-95222, 35-96137, 37-01399, 37-02147, 37-05105,
37-10576, 37-12827, 37-16040, 37-16873, 37-20114, 37-22351,
37-26799, 37-27245, 37-33717, 37-36371, 37-37409, 41-11004,
41-11898, 41-11946, 41-12989;
(c) 21-65779, 21-75172, 32-67143, 32-67287, 35-39374,
35-39472, 35-40820, 35-45866, 35-50970, 35-51753, 35-52477,
35-53952, 35-53986, 35-56697, 35-60827, 35-64630, 35-65208,
35-71771, 35-73412, 35-76793, 35-77105, 35-77285, 35-77492,
35-82167, 35-89389, 35-89831, 35-91751, 35-94573, 35-95222,
35-96137, 35-97931, 37-01399, 37-12827, 37-16040, 37-16873,
37-21703, 37-22351, 37-26799, 37-27245, 37-30921, 37-33717,
37-36838, 41-09170, 41-10558, 41-11004, 41-11898, 41-11946,
41-12982, 41-12989, 41-15625;
(d) 35-34999, 35-42151, 35-47282, 35-55003, 35-59369,
35-87288, 35-02147, 37-05105, 37-10576, 37-20114, 37-36371,
37-37409;
(e) 35-89831, 35-97931, 35-21703, 37-30921, 37-36838,
41-11004, 41-11898, 41-12982.
Manufacturer: Michigan Community Blood Center, doing business as
Grand Valley Blood Program, Grand Rapids, Michigan.
Recalled by: Manufacturer, by telephone November 1, 1990, and by
letters of January 2, 7, and 14, 1991. Firm-initiated
recall ongoing.
Distribution: Mississippi, Virginia, New York, California, New Jersey,
Pennsylvania, Michigan.
Quantity: (a) 62 units; (b) 50 units; (c) 50 units; (d) 12 units;
(e) 8 units.
Reason: Blood components, which tested non-reactive for the
antibody to the human immunodeficiency virus type 1
(anti-HIV-1) but were collected from donors who
previously tested repeatably reactive for anti-HIV-1,
were distributed.
----
-9-
Product: Source Plasma. Recall #B-144-2.
Code: Unit #52396.
Manufacturer: Sera-Tec Biologicals, Inc., Toledo, Ohio.
Recalled by: Consignee notified firm on or about July 10, 1991.
Recall complete.
Distribution: Michigan.
Quantity: 1 unit.
Reason: Blood products from two different donors were
labeled with identical bleed numbers and released.
----
Product: Red Blood Cells. Recall #B-146-2.
Code: Unit #6035521.
Manufacturer: Community Blood Center, Inc., Appleton, Wisconsin.
Recalled by: Consignee notified firm by telephone December 31, 1991.
Recall complete.
Distribution: Kentucky.
Quantity: 1 unit.
Reason: A unit of Red Blood Cells, labeled with the incorrect
expiration date, was distributed.
----
DEVICES AND RADIOLOGICAL PRODUCTS
Class I -
NONE
Class II -
CORRECTION: Recall #Z-295-2, which appeared in the February 12, 1992
Enforcement Report should read:
Product: Externally Assembled Lower Limb Prosthesis, AKA, Above
Knee Prosthesis Appliance/Artificial Leg. Only those
units manufactured using the Teh Lin above the knee wood
block, TWB-4 series are involved. Only units manufactured
from 1/91 through 10/91 may contain this component. This
is a custom device that is labeled: Harry J. Lawall & Son,
Philadelphia, Pennsylvania. Recall #Z-295-2.
Reason: The Teh Lin above the knee wood block, TWB-4 series
component appeared to be fracturing while in use and
may cause a severe malalignment of the leg components.
----
-10-
Class III -
Product: ZMS IM Fixation Tibial Nails, orthopedic implant products
used in fracture fixation of the tibia:
1. Catalog No. 2238-09-40, 9mm diameter and 40cm length
2. Catalog No. 2238-09-42, 9mm diameter and 42cm length
3. Catalog No. 2238-09-44, 9mm diameter and 44cm length
4. Catalog No. 2238-10-40, 10mm diameter and 40cm length
5. Catalog No. 2238-10-42, 10mm diameter and 42cm length
6. Catalog No. 2238-10-44, 10mm diameter and 44cm length
7. Catalog No. 2238-11-40, 11mm diameter and 40cm length
8. Catalog No. 2238-11-42, 11mm diameter and 42cm length
9. Catalog No. 2238-11-44, 11mm diameter and 44cm length
10. Catalog No. 2238-12-40, 12mm diameter and 40cm length
11. Catalog No. 2238-12-42, 12mm diameter and 42cm length
12. Catalog No. 2238-12-44, 12mm diameter and 44cm length
13. Catalog No. 2238-13-40, 13mm diameter and 40cm length
14. Catalog No. 2238-13-42, 13mm diameter and 44cm length
15. Catalog No. 2238-13-44, 13mm diameter and 44cm length
16. Catalog No. 2238-14-40, 14mm diameter and 40cm length
17. Catalog No. 2238-14-42, 14mm diameter and 42cm length
18. Catalog No. 2238-14-44, 14mm diameter and 44cm length
Recall #Z-323/340-2.
Code: Lot numbers:
(1) 78262100; (2) 78262200; (3) 78262300; (4) 77691500;
(5) 77691600; (6) 77691700; (7) 78610600; (8) 78610700;
(9) 78610800; (10) 78734300; (11) 78734200; (12) 78734100;
(13) 77963900; (14) 77964000; (15) 77964100; (16) 7808900;
(17) 708089000; (18) 78089100.
Manufacturer: Zimmer, Inc., Warsaw, Indiana.
Recalled by: Manufacturer, by letter December 23, 1991. Firm-initiated
recall ongoing.
Distribution: Kentucky, New Jersey, Connecticut, Wisconsin, Virginia,
Mississippi, California, South Carolina, Ohio.
Quantity: 76 pieces were distributed.
Reason: The tibial nails have light green stickers that code for
femoral nails instead of the correct black stickers for
tibial nails.
----
VETERINARY PRODUCTS
NONE
-11-