FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
02/26/1992

Recalls and Field Corrections:  February 26, 1992


                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  product will cause serious adverse health consequences or
                  death.


Product:          (a) Frozen Razor Clams in 5 pound cans; (b) Minced Razor
                  Clams, in 7 ounce cans.  Recall #F-116/117-2.
Code:             (a) All lots produced since 9/1/91; (b) Lot #01118/BMRCL.
Manufacturer:     Bell Buoy Crab Company, Seaside, Oregon.
Recalled by:      Manufacturer, by telephone November 21-22, 1991.
                  Firm-initiated recall complete.
Distribution:     Washington state, Oregon, California, Virginia, Idaho.
Quantity:         (a) 591 cans; (b) 96 cans were distributed.
Reason:           Product is contaminated with Domoic Acid.

                                     ----

                                      -1-
                                       


                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the
                  probability of serious adverse health consequences is
                  remote.


Product:          Mehaffies brand (a) Pineapple Pie; (b) Lemon Pies,
                  32 ounces.  Recall #F-120/121-2.
Code:             Pies coded with a pull date that expires 4 days from
                  the date of manufacture.
Manufacturer:     Mehaffies Pie Company, Dayton, Ohio.
Recalled by:      Manufacturer, by visit July 11, 1991.  Firm-initiated
                  recall complete.
Distribution:     Ohio.
Quantity:         Approximately (a) 80 pies; (b) 300 pies were distributed;
                  none remains on market.
Reason:           Product contained undeclared FD&C Yellow No. 5.

                                     ----

Product:          Birds Eye Brussels Sprouts, in 10 ounce foil wrapped
                  boxes.  Recall #F-122-2.
Code:             Lot numbers NI0348 5, NI0348 6, NI0348 7.
Manufacturer:     Norcal Crosetti Foods, Inc., Watsonville, California.
Recalled by:      Kraft General Foods, Inc., Glenview, Illinois, by letter
                  January 23, 1992.  Firm-initiated recall complete.
Distribution:     Nationwide.
Quantity:         1,515 cases (24 boxes per case) were distributed; firm
                  estimates 30 percent remains on the market.
Reason:           Product is contaminated with diesel fuel.

                                     ----

Product:          Kountry Fresh brand (a) Shelled Peanuts; (b) Roasted
                  Peanuts, in 16 ounce glass jars.  Recall #F-123/124-2.
Code:             Sell by 5-5-93B.
Manufacturer:     John B. Sanfilippo & Son, Inc., Bainbridge, Georgia.
Recalled by:      Manufacturer, by visit on or about January 31, 1992.
                  Firm-initiated recall ongoing.
Distribution:     Alabama.
Quantity:         Unknown.
Reason:           Product contains high aflatoxin levels.

                                     ----

Product:          Fresh Asparagus imported from Peru, in unlabeled 11
                  pound cartons.  Recall #F-125-2.
Code:             None.
Manufacturer:     Frutech International Pompnao Beach, Florida.
Recalled by:      Manufacturer, by telephone November 20, 1991, followed
                  by letter November 21, 1991.  Firm-initiated recall
                  complete.

                                      -2-



Distribution:     Florida, Michigan, New York, Georgia.
Quantity:         392 cartons were distributed; firm estimates none
                  remains remain on the market.
Reason:           Product is contaminated with Chlorpyrifos.

                                     ----

Product:          Blended Margarine and Butter Products:
                  (a) TW Services, 1 pound containers;
                  (b) Gold-N-Sweet, 5 pound tubs;
                  (c) Old Fashion, 20 pound pails;
                  (d) One + Two (1+2), 1 pound containers, and 5 and
                  50 pound pails;
                  (e) Fancy Blend, 30 and 50 pound cubes.
                  Recall #F-127/131-2.
Code:             All lot codes that include the letters "CW".
Manufacturer:     Wilsey Foods, Inc., City of Industry, California.
Recalled by:      Manufacturer, by telephone November 12, 1991 and by
                  press release issued by the State of California,
                  Department of Health Services on November 14-15, 1991.
                  Firm-initiated recall ongoing.
Distribution:     Arizona, California, Hawaii, Nevada, Utah.
Quantity:         Approximately 1,660,000 pounds were manufactured;
                  FDA estimates that little product remains in commerce.
Reason:           Product is contaminated with Staphylococcus enterotoxin.

                                     ----

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse
                  health consequences.


CORRECTION:       Bottled Water, Recall #F-112-2, manufactured by
                  Cascade Clear Water Company, Sedro Wooley, Washington,
                  which appeared in the January 29, 1992 Enforcement
                  Report is a Class III recall.

                                     ----

                                   COSMETICS
                  Class I -

                                     NONE
                  Class II -

Product:          Hand Lotion, non-medicated, in 12 fluid ounce plastic
                  pump bottles.  Recall #F-126-2.
Code:             Lot numbers: MG0390, MG1190, MG1290, MG1390, MH0190
                  MH0290, MH0990, MH1090, MH1790, MH2090, MH2190.
Manufacturer:     Midik Corporation, Elkhart, Indiana.

                                      -3-



Recalled by:      Bristol-Myers Squibb Company, Evansville, Indiana, by
                  voice mail message on October 25, 1991, followed by
                  sales department bulletin dated October 30, 1991, and
                  by letter October 31, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         84,660 bottles were distributed; firm estimates none
                  remains on the market.
Reason:           Product is contaminated with Pseudomonas Aeruginosa.

                                     ----

                  Class III -
                                     NONE

                           HUMAN DRUGS AND BIOLOGICS

                  Class I -

                                     NONE
                  Class II - 

Product:          LyphoMed brand Nitroglycerin Injection, USP, 5 mg/ml, in
                  5 ml and 10 ml vials, a Rx injectable for intravenous
                  infusion after dilution for control of blood pressure
                  during surgical procedures.  Recall #D-214-2.
Code:             Lot numbers:  110359 EXP 10/92 (5 ml), 110367 EXP 10/92
                  (10 ml).
Manufacturer:     Lyphomed, Division of Fujisawa USA, Inc., Melrose Park,
                  Illinois.
Recalled by:      Manufacturer, by letter February 7, 1992.  Firm-initiated
                  recall ongoing.
Distribution:     Florida, New York, California, New Jersey, Puerto Rico,
                  India.
Quantity:         24,250 5-ml vials and 15,720 10-ml vials were distributed;
                  firm estimates none remains on the market.
Reason:           Subpotency.
                                     ----

Product:          Levothroid (Levothyroxin Sodium, USP) Tablets, 100 mcg,
                  in 100 tablet bottles, a Rx product used as a replacement
                  for substitution therapy for diminished or absent thyroid
                  function.  Recall #D-216-2.
Code:             05428-1 EXP 4/93.
Manufacturer:     Pharmacy Dispensing Quantities, Inc., Fort Scott, Kansas.
                  (repacker/responsible firm).
Recalled by:      Repacker, by letter December 23, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide.
Quantity:         1,498 bottles were distributed.
Reason:           An Inderal 10 mg tablet was found in a bottle of the
                  product.

                                     ----

                                      -4-



Product:          OTC Acetaminophen Liquid, cherry and grape flavors, in
                  4 fluid ounce plastic bottles, with accompanying dosage
                  cup, indicated for relief of children's fever and pain,
                  under the K.C. Pharmaceuticals, Leiner Nutritional
                  Products, and Topco Associates labels.  Recall #D-219-2.
Code:             Lot numbers 9B21401, 9C21402, 9F21403, 9F21404, 0A025,
                  0E048, OK018, 0M016, 0N029, 1D033, 1D035, EXP dates range
                  from 3/91 to 11/92.
Manufacturer:     K.C. Pharmaceuticals, Inc., Pomona, California.
Recalled by:      Manufacturer, by letter January 10, 1992.  Firm-initiated
                  recall ongoing.
Distribution:     California, Illinois.
Quantity:         81,822 bottles were distributed; FDA estimates 8,000
                  bottles remain on the market.
Reason:           Markings on dosage cup not totally compatible with label
                  dosage instructions.

                                     ----

Product:          Platelets Pheresis.  Recall #B-121-2.
Code:             Unit numbers:  21P27268, 21P27270, 21P27273S, 21P27274S,
                  21P27274T.
Manufacturer:     American Red Cross, Portland, Oregon.
Recalled by:      Manufacturer, by telephone June 6, 1991, followed by
                  letter June 21, 1991.  Firm-initiated recall complete.
Distribution:     Oregon.
Quantity:         5 units.
Reason:           Blood products, labeled with incorrect expiration dates,
                  were distributed.

                                     ----

Product:          Recovered Plasma.  Recall #B-128-2.
Code:             Unit #53R77805.
Manufacturer:     American Red Cross Blood Services, Baltimore, Maryland.
Recalled by:      Manufacturer, by letter June 26, 1991.  Firm-initiated
                  recall complete.
Distribution:     Switzerland.
Quantity:         1 unit.
Reason:           Blood product drawn from an unsuitable donor was
                  distributed.

                                     ----

Product:          Red Blood Cells.  Recall #B-130-2.
Code:             Unit #1390990.
Manufacturer:     Mississippi Blood Services, Inc., Jackson, Mississippi.
Recalled by:      Manufacturer, by telephone October 17, 1991, followed
                  by letter November 14, 1991.  Firm-initiated recall
                  complete.
Distribution:     Mississippi.

                                      -5-


Quantity:         1 unit.
Reason:           A unit of red blood cells, testing repeatably reactive
                  for the antibody to hepatitis C virus encoded antigen
                  (anti-HCV), was distributed.

                                     ----

Product:          Red Blood Cells.  Recall #B-131-2.
Code:             Unit #1678135.
Manufacturer:     Walson Army Hospital, Ft. Dix, New Jersey.
Recalled by:      Manufacturer, by telephone September 26 or 27, 1991.
                  Firm-initiated recall complete.
Distribution:     New York.
Quantity:         1 unit.
Reason:           Blood component which tested repeatably reactive for
                  hepatitis B surface antigen (HBsAg) was distributed for
                  transfusion.
                                     ----

Product:          (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen
                  Plasma; (d) Recovered Plasma.  Recall #B-135/138-2.
Code:             Unit numbers:  21-54448, 21-58709, 21-62534, 21-64150,
                  21-66371, 21-68787, 21-69882, 21-75061, 21-85046, 21-87455,
                  21-90189, 35-17567, 35-20290, 35-22115, 35-25034, 35-37562,
                  35-39564, 35-45190, 35-57740, 35-58126, 35-69974, 35-72033,
                  35-77937, 35-78390, 35-79838, 35-87387, 35-94708, 35-99103,
                  37-01509, 37-01906, 37-05957, 37-06798, 37-13446, 37-24155,
                  37-29406, 37-31288;
                  (b) 35-17567, 35-20290, 35-22115, 35-39564, 35-57740,
                  35-58126, 35-69974, 35-72033, 35-79838, 35-78390, 35-87387,
                  35-94708, 35-99103, 37-01509, 37-01906, 37-05957, 37-06798,
                  37-13446, 37-24155, 37-29406, 37-31288;
                  (c) 21-85046, 35-20290, 35-37562, 35-76782, 37-01906;
                  (d) 21-54448, 21-58709, 21-62534, 21-64150, 21-66371,
                  21-68787, 21-69882, 21-75061, 21-87455, 21-90189, 35-17567,
                  35-22115, 35-25034, 35-39564, 35-45190, 35-57740, 35-58126,
                  35-69974, 35-72033, 35-77937, 35-78390, 35-79838, 35-87387,
                  35-94708, 35-99103, 37-01509, 37-05957, 37-06798, 37-13446,
                  37-24155, 37-29406, 37-31288.
Manufacturer:     Michigan Community Blood Center, doing business as
                  Grand Valley Blood Program, Grand Rapids, Michigan.
Recalled by:      Manufacturer, by telephone November 1, 1990, and by
                  letters of January 2, 7 and 14, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Mississippi, Virginia, New York, California, New Jersey,
                  Pennsylvania, Michigan.
Quantity:         (a) 36 units; (b) 21 units; (c) 5 units; (d) 32 units.
Reason:           Blood components, which tested non-reactive for the
                  antibody to the human immunodeficiency virus type 1
                  (anti-HIV-1) but were collected from donors who
                  previously tested repeatably reactive for anti-HIV-1,
                  were distributed.
                                     ----

                                      -6-


Product:          Source Plasma.  Recall #B-145-2.
Code:             Unit #XF9880.
Manufacturer:     Indianapolis Blood Plasma, Inc., Indianapolis, Indiana.
Recalled by:      Consignee notified firm October 8, 1991.  Recall complete.
Distribution:     North Carolina.
Quantity:         1 unit.
Reason:           Blood product untested for the antibody to the human
                  immunodeficiency virus type 1 (anti-HIV-1) and hepatitis B
                  surface antigen (HBsAg) was distributed.

                                     ----

Product:          Platelets, Pooled.  Recall #B-147-2.
Code:             Pool #FF63382.
Manufacturer:     The Blood Center of Southeastern Wisconsin, Milwaukee,
                  Wisconsin.
Recalled by:      Manufacturer, by telephone July 25, 1991.  Firm-initiated
                  recall complete.
Distribution:     Wisconsin.
Quantity:         1 unit.
Reason:           Pooled Platelets, which contained platelets collected
                  from an ineligible donor due to a reported history of
                  hepatitis b exposure occurring two weeks prior to
                  donation, were distributed.

                                     ----

                  Class III -

Product:          Rx Tylenol with Codeine No. 3 Tablets, in 100 tablet
                  bottles.  Recall #D-215-2.
Code:             Lot #HC1632P EXP 3/94.
Manufacturer:     McNeil Pharmaceutical Company, Dorado, Puerto Rico.
Recalled by:      McNeil Pharmaceutical, Spring House, Pennsylvania,
                  by letter January 20, 1992.  Firm-initiated recall ongoing.
Distribution:     New York, Pennsylvania, Florida, Virginia, Maryland,
                  Georgia, Alabama, West Virginia, South Carolina, Rhode
                  Island, Connecticut, North Carolina, Maine, Vermont.
Quantity:         13,680 bottles were distributed; firm estimates 1,000
                  bottles remain on the market.
Reason:           Some bottles in correctly labeled unit cartons were
                  placed in a packer labeled Tylenol with Codeine No. 2
                  tablets.  The outer shipping carton is correctly labeled.

                                     ----

Product:          Trandate (Labetalol Hydrochloride) Tablets, 100 mg, in
                  100 tablet bottles, a Rx drug for use in the management
                  of hypertension.  Recall #D-217-2.
Code:             Z60300MX EXP MAY94.
Manufacturer:     Pharmacy Dispensing Quantities, Inc., Ft. Scott, Kansas
                  (repacker/responsible firm).

                                      -7-



Recalled by:      Repacker, by letter August 13, 1991.  Firm-initiated
                  recall complete.
Distribution:     New York, Missouri, Tennessee, South Dakota, Idaho,
                  Ohio, Pennsylvania, Indiana.
Quantity:         140 bottles were distributed.
Reason:           The label correctly declares the product as 100 mg
                  but the ingredient statement shows the product as 200 mg.

                                     ----

Product:          Clonidine HCl Tablets, 0.3 mg, in 100 tablet bottles, a
                  Rx drug indicated for the treatment of hypertension,
                  under the Best label only.  Recall #D-218-2.
Code:             Lot numbers 322991 EXP 8/93, 127491 EXP 9/93.
Manufacturer:     Par Pharmaceutical, Inc., Spring Valley, New York.
Recalled by:      Manufacturer.  Firm-initiated recall ongoing.
Distribution:     Florida.
Quantity:         72 bottles were distributed.
Reason:           Correct tablet number 112 was misprinted as 110 on the
                  side panel of the Best brand label.

                                     ----

Product:          (a) Platelets Pheresis; (b) Source Leukocytes.
                  Recall #B-119/120-2.
Code:             Unit numbers:  (a) 21P27407, 21P27409; (b) 21F12431.
Manufacturer:     American Red Cross, Pacific Northwest Services,
                  Portland, Oregon.
Recalled by:      Manufacturer, by letter June 24, 1991.  Firm-initiated
                  recall complete.
Distribution:     Oregon, Washington state.
Quantity:         (a) 2 units; (b) 1 unit.
Reason:           Blood products, incorrectly tested for the antibody
                  to the hepatitis C virus (anti-HCV), were distributed.

                                     ----

Product:          Red Blood Cells.  Recall #B-129-2.
Code:             Unit numbers: 5161923 through 5161930, 5161932, 5161933,
                  5161935, 5161936, 5161939, 5161944, 5161946, 5161949,
                  5161950, 5161951.
Manufacturer:     Department of the Navy, Naval Hospital Camp Lejeune,
                  Camp Lejeune, North Carolina.
Recalled by:      Manufacturer, by telephone August 28-29, 1991.  Firm-
                  initiated recall complete.
Distribution:     North Carolina, New Jersey.
Quantity:         18 units.
Reason:           Red Blood Cells labeled with incorrect expiration
                  dates were distributed.

                                     ----

                                      -8-



Product:          (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma;
                  (d) Fresh Frozen Plasma; (e) Cryoprecipitated AHF.
                  Recall #B-139/143-2.
Code:             Unit numbers:  (a) 21-65779, 21-75172, 32-67143, 35-34999,
                  35-39374, 35-39472, 35-40820, 35-42151, 35-45886, 35-47282,
                  35-50970, 35-51753, 35-52477, 35-53952, 35-53986, 35-55003,
                  35-56697, 35-59369, 35-60827, 35-64630, 35-65208, 35-71771,
                  35-73412, 35-76782, 35-76793, 35-77105, 35-77285, 35-77492,
                  35-82167, 35-87288, 35-89389, 35-89831, 35-91751, 35-94573,
                  35-95222, 35-96137, 35-97931, 35-01399, 37-02147, 37-05105,
                  37-10576, 37-12827, 37-16040, 37-16873, 37-20114, 37-21703,
                  37-22351, 37-26799, 37-27245, 37-30921, 37-33717, 37-36371,
                  37-36838, 37-37409, 41-09170, 41-10558, 41-11004, 41-11898,
                  41-11946, 41-12982, 41-12989, 41-15625;
                  (b) 32-67143, 35-34999, 35-39374, 35-39472, 35-40820,
                  35-42151, 35-45886, 35-47282, 35-50970, 35-51753, 35-52477,
                  35-53952, 35-53986, 35-55003, 35-56697, 35-59369, 35-60827,
                  35-64630, 35-65208, 35-71771, 35-73412, 35-76782, 35-76793,
                  35-77105, 35-77285, 35-82167, 35-87288, 35-89389, 35-91751,
                  35-94573, 35-95222, 35-96137, 37-01399, 37-02147, 37-05105,
                  37-10576, 37-12827, 37-16040, 37-16873, 37-20114, 37-22351,
                  37-26799, 37-27245, 37-33717, 37-36371, 37-37409, 41-11004,
                  41-11898, 41-11946, 41-12989;
                  (c) 21-65779, 21-75172, 32-67143, 32-67287, 35-39374,
                  35-39472, 35-40820, 35-45866, 35-50970, 35-51753, 35-52477,
                  35-53952, 35-53986, 35-56697, 35-60827, 35-64630, 35-65208,
                  35-71771, 35-73412, 35-76793, 35-77105, 35-77285, 35-77492,
                  35-82167, 35-89389, 35-89831, 35-91751, 35-94573, 35-95222,
                  35-96137, 35-97931, 37-01399, 37-12827, 37-16040, 37-16873,
                  37-21703, 37-22351, 37-26799, 37-27245, 37-30921, 37-33717,
                  37-36838, 41-09170, 41-10558, 41-11004, 41-11898, 41-11946,
                  41-12982, 41-12989, 41-15625;
                  (d) 35-34999, 35-42151, 35-47282, 35-55003, 35-59369,
                  35-87288, 35-02147, 37-05105, 37-10576, 37-20114, 37-36371,
                  37-37409;
                  (e) 35-89831, 35-97931, 35-21703, 37-30921, 37-36838,
                  41-11004, 41-11898, 41-12982.
Manufacturer:     Michigan Community Blood Center, doing business as
                  Grand Valley Blood Program, Grand Rapids, Michigan.
Recalled by:      Manufacturer, by telephone November 1, 1990, and by
                  letters of January 2, 7, and 14, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Mississippi, Virginia, New York, California, New Jersey,
                  Pennsylvania, Michigan.
Quantity:         (a) 62 units; (b) 50 units; (c) 50 units; (d) 12 units;
                  (e) 8 units.
Reason:           Blood components, which tested non-reactive for the
                  antibody to the human immunodeficiency virus type 1
                  (anti-HIV-1) but were collected from donors who
                  previously tested repeatably reactive for anti-HIV-1,
                  were distributed.

                                     ----

                                      -9-



Product:          Source Plasma.  Recall #B-144-2.
Code:             Unit #52396.
Manufacturer:     Sera-Tec Biologicals, Inc., Toledo, Ohio.
Recalled by:      Consignee notified firm on or about July 10, 1991.
                  Recall complete.
Distribution:     Michigan.
Quantity:         1 unit.
Reason:           Blood products from two different donors were
                  labeled with identical bleed numbers and released.

                                     ----

Product:          Red Blood Cells.  Recall #B-146-2.
Code:             Unit #6035521.
Manufacturer:     Community Blood Center, Inc., Appleton, Wisconsin.
Recalled by:      Consignee notified firm by telephone December 31, 1991.
                  Recall complete.
Distribution:     Kentucky.
Quantity:         1 unit.
Reason:           A unit of Red Blood Cells, labeled with the incorrect
                  expiration date, was distributed.

                                     ----

                       DEVICES AND RADIOLOGICAL PRODUCTS


                  Class I -

                                     NONE

                  Class II -

CORRECTION:       Recall #Z-295-2, which appeared in the February 12, 1992
                  Enforcement Report should read:

Product:          Externally Assembled Lower Limb Prosthesis, AKA, Above 
                  Knee Prosthesis Appliance/Artificial Leg.  Only those
                  units manufactured using the Teh Lin above the knee wood
                  block, TWB-4 series are involved.  Only units manufactured
                  from 1/91 through 10/91 may contain this component.  This
                  is a custom device that is labeled:  Harry J. Lawall & Son,
                  Philadelphia, Pennsylvania.  Recall #Z-295-2.

Reason:           The Teh Lin above the knee wood block, TWB-4 series
                  component appeared to be fracturing while in use and
                  may cause a severe malalignment of the leg components.

                                     ----

                                     -10-



                  Class III -

Product:          ZMS IM Fixation Tibial Nails, orthopedic implant products
                  used in fracture fixation of the tibia:
                  1.  Catalog No. 2238-09-40, 9mm diameter and 40cm length
                  2.  Catalog No. 2238-09-42, 9mm diameter and 42cm length
                  3.  Catalog No. 2238-09-44, 9mm diameter and 44cm length
                  4.  Catalog No. 2238-10-40, 10mm diameter and 40cm length
                  5.  Catalog No. 2238-10-42, 10mm diameter and 42cm length
                  6.  Catalog No. 2238-10-44, 10mm diameter and 44cm length
                  7.  Catalog No. 2238-11-40, 11mm diameter and 40cm length
                  8.  Catalog No. 2238-11-42, 11mm diameter and 42cm length
                  9.  Catalog No. 2238-11-44, 11mm diameter and 44cm length
                  10. Catalog No. 2238-12-40, 12mm diameter and 40cm length
                  11. Catalog No. 2238-12-42, 12mm diameter and 42cm length
                  12. Catalog No. 2238-12-44, 12mm diameter and 44cm length
                  13. Catalog No. 2238-13-40, 13mm diameter and 40cm length
                  14. Catalog No. 2238-13-42, 13mm diameter and 44cm length
                  15. Catalog No. 2238-13-44, 13mm diameter and 44cm length
                  16. Catalog No. 2238-14-40, 14mm diameter and 40cm length
                  17. Catalog No. 2238-14-42, 14mm diameter and 42cm length
                  18. Catalog No. 2238-14-44, 14mm diameter and 44cm length
                  Recall #Z-323/340-2.
Code:             Lot numbers:
                  (1) 78262100; (2) 78262200; (3) 78262300; (4) 77691500;
                  (5) 77691600; (6) 77691700; (7) 78610600; (8) 78610700;
                  (9) 78610800; (10) 78734300; (11) 78734200; (12) 78734100;
                  (13) 77963900; (14) 77964000; (15) 77964100; (16) 7808900;
                  (17) 708089000; (18) 78089100.
Manufacturer:     Zimmer, Inc., Warsaw, Indiana.
Recalled by:      Manufacturer, by letter December 23, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Kentucky, New Jersey, Connecticut, Wisconsin, Virginia,
                  Mississippi, California, South Carolina, Ohio.
Quantity:         76 pieces were distributed.
Reason:           The tibial nails have light green stickers that code for
                  femoral nails instead of the correct black stickers for
                  tibial nails.
                                     ----

                             VETERINARY PRODUCTS 

                                     NONE

                                     -11-