11/20/1991

Recalls and Field Corrections: November 20, 1991

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or
                  death.

                                     NONE

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the
                  probability of serious adverse health consequences is
                  remote.

                                     NONE

                                      -1-
                                       



                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse
                  health consequences.


Product:          Individually Quick Frozen Slipper/Lobster, in 5 pound
                  plastic bags, 6 bags per shipping carton:
                  (a) Lobster tails, 4-6 ounce size;
                  (b) Lobster Tail Meat, 1-3 ounce size.  Recall #F-044/045-2.
Code:             Lot numbers: (a) 47900; (b) 47901.
Manufacturer:     Sun Wah Marine Products (Far East) Company, Ltd.,
                  Tun Mun, N.T. Hong Kong.
Recalled by:      Central Seaway Company, Inc., Northfield, Illinois, by
                  telephone September 9, 1991 followed by letter
                  September 24, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         (a) 146 cartons; (b) 715 cartons were distributed;
                  firm estimates that very little remains on the market.
Reason:           Product is overglazed (contains excessive ice) and tails
                  are decomposed.
                                     ----

Product:          Dyna-Bor Plus, in 1 ounce bottles, a dietary supplement
                  formulated as a companion to all types of exercise.
                  Recall #F-046-2.
Code:             Lot #127880.
Manufacturer:     Vita Fore Products Company, Ozone Park, New York.
Recalled by:      Dynamic Power Labs, Inc., Wheeler, Indiana, by letter
                  July 16, 1991.  Firm-initiated recall complete.
Distribution:     Indiana, California, New Jersey, Minnesota.
Quantity:         5,520 bottles were distributed; firm estimates less
                  than 1 percent of the product remains on the market.
Reason:           Product is deficient in niacin.

                                     ----

                                   COSMETICS

                                     NONE

                           HUMAN DRUGS AND BIOLOGICS

                  Class I -
                                     NONE
                  Class II - 

Product:          Perphenazine and Amitriptyline HCl Tablets, USP 4 mg/10 mg,
                  in 100 and 500 tablet bottles, a Rx psychotherapeutic agent,
                  under the following labels:  Mason, Zenith, Schein, Bioline,
                  Goldline, and H.L. Moore.  Recall #D-037-2.
Code:             Zenith lot #2938-06 EXP 11/91.
Manufacturer:     Zenith Laboratories, Inc., Northvale, New Jersey.

                                      -2-



Recalled by:      Manufacturer, by letters dated September 6, 1991.
                  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         Approximately 928,500 tablets were distributed; firm
                  estimates none remains on the market.
Reason:           Subpotency.

                                     ----

Product:          Fifty-one (51) prescription drug pharmaceutical
                  products (the total number also includes the number of
                  strengths for those products with more than one strength);
                  Recall #D-038/088-2.

(a)  Acetohexamide, USP, 250 mg & 500 mg Tablets, used as an oral
blood-glucose lowering drug, packaged in bottles of 100 and 200 tablets under
the following labels:  Aligen, Best, Bioline, Geneva Generics,
Goldline, PBI, and Rugby.

(b)  Baclofen, USP, 10 mg & 20 mg Tablets, used as muscle relaxant, and
packaged in bottles of 100 and 250 tablets under the following labels:
Bioline, Geneva Generics, Goldline, Major, PARMED, PBI, Rugby, Schein, and URL.

(c)  Benztropine Mesylate, USP, 0.5 mg, 1.0 mg, & 2.0 mg Tablets, used as
an adjunct in therapy of all forms of parkinsonism, and packaged in bottles
of 100, 500, & 1000 tablets under the following labels:  Aligen, Best, Geneva
Generics, Major, UDL Laboratories, PARMED, PBI, and Rugby.

(d)  Chlordiazepoxide and Amitriptyline Hydrochloride, USP, 5 mg/12.5 mg &
10 mg/25 mg Tablets, indicated for treatment of patients with moderate to
severe depression associated with moderate to severe anxiety, and packaged
in bottles of 100 & 500 tablets under the following labels: Aligen, Geneva      
Generics, Major, MARTEC, PBI, and URL.

(e)  Chlorpropamide Tablets, USP, 100 mg & 250 mg Tablets, an oral
blood-glucose lowering drug, and packaged in bottles of 100, 500, & 1000
tablets under the following labels:  Best, Lemmon, PBI, and URL.

(f)  Desipramine Hydrochloride, USP, 25 mg, 50 mg, 75 mg, & 100 mg Tablets,
an antidepressant drug indicated for relief of symptoms in various depressive
syndromes, and packaged in bottles of 100 & 1000 tablets under the following
labels:  Aligen, Best, Bioline, Goldline, Lemmon, Major, MARTEC, H.L. Moore,
PARMED, PBI, Rugby, Schein, and URL.

(g)  Dipyridamole, USP, 25 mg, 50 mg, & 75 mg Tablets, is classified as a
coronary vasodilator and is packaged in bottles of 100 & 1000 tablets
under the PBI label.

(h)  Flurazepam Hydrochloride, USP, 15 mg & 30 mg Capsules, a hypnotic agent
indicated for use as a treatment of insomnia characterized by difficulty in
falling asleep, and packaged in bottles of 100 & 500 capsules under the
following labels:  MARTEC, PBI, and URL.

                                      -3-



(i)  Hydrocodone Bitartrate/APAP, 5 mg/500 mg Tablets, indicated for use in
relief of moderate to moderately severe pain, and packaged in bottles of 100
& 500 tablets under the following labels:  Best, Bioline, Geneva Generics,
Goldline, DuoCet, H.L. Moore, PARMED, PBI, Rugby, and URL.

(j)  Hydroxyzine Hydrochloride, USP, 10 mg, 25 mg, & 50 mg Tablets, is an
anti-psychotic drug used as a tranquilizer, and packaged in bottles of 100,
500, & 1000 tablets under the following labels: Best, PARMED, PBI, and Rugby.

(k)  Lorazepam, USP, 1 mg & 2 mg Tablets, an antianxiety agent indicated
for use in the management of anxiety disorders or for the short-term relief
of symptoms of anxiety, and packaged in bottles of 100, 500, & 1000 tablets
under the following labels:  Best & PBI.

(l)  Megestrol Acetate, USP, 20 mg & 40 mg Tablets, indicated for the
palliative treatment of advanced carcinoma of the breast or endometrium
(i.e., recurrent, inoperable, or metastatic disease), and packaged
in bottles of 100, 250, & 500 tablets under the following labels:  Bioline,
Geneva Generics, Goldline, Major, Mason, H.L. Moore, PARMED, PBI, Qualitest,
Rugby, Schein, and URL.

(m)  Metaproterenol Sulfate, USP, 10 mg & 20 mg Tablets, indicated for use
as a bronchodilator for bronchial asthma, and for reversible bronchospasm
which may occur in association with bronchitis and emphysema, and packaged
in bottles of 100 tablets under the following labels: Best, Bioline, Genetco,
Geneva Generics, Goldline, Major, H.L. Moore, PARMED, PBI, Rugby, Schein, and
URL.

(n)  Methyclothiazide, USP, 5 mg Tablets, indicated for use in the management
of hypertension, and packaged in bottles of 100 & 1000 tablets under the
following labels:  Aligen, Best, Bioline, Goldline, Major, Mason, H.L. Moore,
PBI, Qualitest, Rugby, Schein, and URL.

(o)  Metoclopramide Hydrochloride, USP, 10 mg Tablets, indicated for use as
an anti-emetic/nauseant drug, and packaged in bottles of 100, 500, &
1000 tablets under the PBI and URL labels.

(p)  Minoxidil, USP, 2.5 mg Tablets, indicated for use in treatment of
hypertension, and packaged in bottles of 100 tablets under the
following labels:  Major, PBI, and Rugby.

(q)  Oxybutynin Chloride, USP, 5 mg Tablets, indicated for use in
relief of symptoms of neurogenic bladder instability, and packaged in bottles
of 100 & 1000 tablets under the following labels:  Best, Bioline, Genetco,
Geneva Generics, Goldline, Major, Mason, H.L. Moore, PARMED, PBI, Rugby,
Schein, and URL.

(r)  Prazepam, USP, 5 mg & 10 mg Capsules, indicated for the management of
anxiety disorders, and packaged in bottles of 100 & 500 capsules under the
following labels:  Aligen, Best, Bioline, Geneva Generics, Goldline, Major,
MARTEC, Mason, H.L. Moore, PARMED, PBI, Qualitest, Rugby, Schein, and URL.


                                      -4-




(s)  Reserpine 0.125 mg/Hydroflumethiazide 50 mg Tablets, anti-hypertensive
drug indicated for use in lowering blood pressure, and packaged in bottles of
100, 500, & 1000 tablets under the PBI label.

(t)  Temazepam, 15 mg & 30 mg Capsules, indicated for use in the relief of
insomnia, and packaged in bottles of 100 & 500 capsules under the following
labels:  Best, Major, MARTEC, PARMED, and PBI.

(u)  Timolol Maleate, USP, 5 mg, 10 mg, & 20 mg Tablets, indicated for use
in treatment of hypertension, and packaged in bottles of 100 tablets under the
following labels:  Best, Bioline, Goldline, Major, Mason, H.L. Moore, PARMED,
PBI, Qualitest, Rugby, Schein, and URL.

(v)  Tolazamide, USP, 250 mg, & 500 mg Tablets, indicated as an adjunct to
diet to lower blood glucose in patients, and packaged in bottles of 100 & 500
tablets under the following labels: MARTEC, PARMED, and PBI.

(w)  Trazodone Hydrochloride, 50 mg & 100 mg Tablets, indicated for use in
treatment of depression, and packaged in bottles of 100, 500 & 1000 tablets
under the following labels:  PARMED and PBI.

(x)  Trimipramine Maleate, 25 mg, 50 mg, & 100 mg Capsules, indicated for
use in symptoms of depression, and packaged in bottles of 100 capsules under
the following labels:  Best, Geneva Generics, H.L. Moore, PBI, Qualitest,
Rugby, Schein, and URL.

(y)  Warfarin Sodium, USP, 2 mg, 2.5 mg, & 5 mg Tablets, anticoagulant, and
packaged in bottles of 100, 250, & 1000 tablets under the following labels:
Aligen, Best, Genetco, Harber, Lemmon, Major, MARTEC, H.L. Moore, PBI,
Qualitest, Rugby, and Schein.
Code:             All lots of affected products manufactured by PBI.
Manufacturer:     Pharmaceutical Basics, Inc., (PBI), Denver, Colorado
Recalled by:      Manufacturer, by letter November 12, 1991.  FDA requested
                  recall ongoing.
Distribution:     Nationwide, Holland, Istanbul, Phillipines, Thailand.
Quantity:         Firm estimates 10-15 percent of products remain on
                  the market.
Reason:           Lack of assurance of bioequivalency and Abbreviated
                  New Drug Application discrepancies.

                                     ----

                  Class III -

Product:          Red Blood Cells.  Recall #B-040-2.
Code:             Unit #5554991.
Manufacturer:     Naval Hospital Blood Bank, Jacksonville, Florida.
Recalled by:      Manufacturer, by telephone January 22, 1991.  Firm-initiated
                  recall complete.
Distribution:     New Jersey.
Quantity:         1 unit.

                                      -5-



Reason:           Blood product, which tested initially reactive for
                  human immunodeficiency virus type 1 (anti-HIV-1),
                  was distributed.

                                     ----

Product:          Red Blood Cells.  Recall #B-041-2.
Code:             Unit #4611334.
Manufacturer:     Central Florida Blood Bank, Orlando, Florida.
Recalled by:      Manufacturer, by telephone and by letter dated June 6,
                  1991.  Firm-initiated field correction (relabeling)
                  complete.
Distribution:     Florida.
Quantity:         1 unit.
Reason:           Blood component was distributed which was labeled
                  with an incorrect expiration date.

                                     ----

Product:          (a) Red Blood Cells; (b) Platelets.  Recall #B-043/044-2.
Code:             Unit numbers: (a) 21G10048, 21G10064, 21G10074
                  21G10102, 21G10118, 21G10119, 21G10120, 21G10123,
                  21G10124, 21G10128, 21G10129, 21G10132, 21G10134,
                  21G10138, 21G10145, 21G10146, 21G10147, 21G10148,
                  21G10150, 21G10152, 21G10158, 21K07242, 21K07258,
                  21K07284, 21M72705, 21M72707, 21M72715, 21Q16167,
                  21Q16168, 21Q16174, 21Q16177, 21Q16178, 21Q16184,
                  21Q16186, 21Q16187, 21Q16189, 21Q16190, 21Q16195,
                  21Q16196, 21Q16197, 21Q16199, 21Q16200;
                  (b) 21M72699, 21M72700, 21M72701, 21M72704, 21M72706,
                  21M72710, 21M72712, 21M72715, 21M72717, 21M72719.
Manufacturer:     American Red Cross Services, Portland, Oregon.
Recalled by:      Manufacturer, by telephone May 22, 1991, followed by
                  letter May 24, 1991.  Firm-initiated recall ongoing.
Distribution:     Washington state, West Virginia, Oregon.
Quantity:         (a) 42 units; (b) 10 units.
Reason:           Blood products, incorrectly tested for the antibody to
                  hepatitis C virus (anti-HCV), were distributed.

                                     ----

Product:          Platelets.  Recall #B-045-2.
Code:             Unit #M43248.
Manufacturer:     Tacoma-Pierce County Hospital, Tacoma, Washington.
Recalled by:      Manufacturer, by telephone June 28, 1991.  Firm-initiated
                  recall complete.
Distribution:     Washington state.
Quantity:         1 unit.
Reason:           Blood component, which was not tested for the antibody
                  to the hepatitis C virus (anti-HCV), was distributed.

                                     ----

                                      -6-




Product:          Platelets.  Recall #B-048-2.
Code:             Unit numbers:  20FF14323, 20FF14325, 20FF14327,
                  20FF14337, 20FF14342, 20FF14343, 20GG20047, 20GG20049,
                  20GG20050, 20GG20051, 20GG20058, 20GG20065, 20GG20073,
                  20GG20074, 20GG20079, 20GG20085, 20GG20089, 20GG20090,
                  20GG20091, 20GG20095, 20GG20102, 20GG20103, 20GG20108,
                  20GG20109, 20GG20111.
Manufacturer:     American Red Cross, Boise, Idaho.
Recalled by:      Manufacturer, by telephone August 15, 1991.  Firm-
                  initiated recall complete.
Distribution:     Idaho.
Quantity:         25 units.
Reason:           Blood products, labeled with incorrect anticoagulant/
                  preservative solution, were distributed.

                                     ----

                       DEVICES AND RADIOLOGICAL PRODUCTS

                  Class I -

EXTENSION NOTE:   Ohmeda Logic 07 Ventilators, Non-rebreathing valves
                  and Black Rubber Membranes, Part No. 6750-0009-200,
                  which appeared in the July 3, 1991 Enforcement Report,
                  has been extended to include all units.

                                     ----

                  Class II -

Product:          Neonatal Monitor (part of Athena Infant Monitoring
                  System), Part #9040.  Recall #Z-046-2.
Code:             Serial numbers:  Omaha Childrens Hospital -- 10293393,
                  10294478, 10294670, 10294671, 10294672, 10294758,
                  10294760, 10294765; South Shore Hospital -- 10294667,
                  10294668, 10294669, 10294771.
Manufacturer:     S & W Medico Teknik, Albertslund, Denmark.
Recalled by:      Air-Shields, Inc., Hatboro, Pennsylvania, by visit
                  September 20, 1991 and October 25, 1991.  Firm-initiated
                  field correction complete.
Distribution:     Nebraska, Massachusetts.
Quantity:         12 units were distributed and have been corrected.
Reason:           The device may fail to monitor or alarm.

                                     ----

Product:          Angestat Introducer with Hemostasis Valve, used to
                  facilitate the percutaneous introduction of catheters
                  or leads into the Vascular System during arterial and 
                  venous procedures:
                  (a) 6 French size, Model #HI6-0 and HI6-2;

                                      -7-


                  (b) 8 French size, Model #HI8-0 and HI8-2;
                  (c) 9 French size, Model #HI9-0 and HI9-2.
                  Recall #Z-078/080-2.
Code:             Lot numbers:  (a) OHBC05, OHBC28, OHBC45, OHBC55, OHBC60,
                  OIBC03, OIBC21, OIBC33, OIBC60, OIBC64, OIBC69, OIBC86,
                  OIBC95, OIBC98, OJBC07, IIBC47;
                  (b) OABC09, OBBC50, ODBC54, OFBC26, OHBC31, OHBC42,
                  OHBC47, OHBC57, OHBC80, OHBC89, OHBC90, OIBC30, OIBC42,
                  OIBC71, OJBC39, OJBC63, OKBC86;
                  (c) OHBC66, OHBC67, OIBC12, OIBC93, OJBC27, OJBC53.
Manufacturer:     Angeion Corporation, Plymouth, Minnesota.
Recalled by:      Manufacturer, by letter September 13, 1991.  Firm-
                  initiated recall ongoing.
Distribution:     Nationwide, Turkey, Germany, England, Singapore.
Quantity:         (a) 5,920 units; (b) 4,891; (c) 1,760 units were
                  distributed.
Reason:           The hemostasis valve can be pulled out of the hub
                  while the catheter is in the blood vessel, when the
                  introducing dilator and guidewire are withdrawn from
                  the vasculature.

                                     ----

Product:          MC II Model #7300E Microcurrent Stimulator - T.E.N.S.
                  (Transcutaneous Electrical Nerve Stimulation) unit,
                  a pain control device.  Recall #Z-081-2.
Code:             Serial number range: 91050001C through 91090060C.
Manufacturer:     Medical Devices, Inc., St. Paul, Minnesota.
Recalled by:      Manufacturer, by letter October 11, 1991.  Firm-initiated
                  field correction ongoing.
Distribution:     Nationwide, Canada.
Quantity:         219 units were distributed.
Reason:           The design of the MC II allows it to send current through
                  the ground wire of the GV II or another MC II unit when
                  they are used in the same Plexus III battery power supply.
                  This allows current to flow in excess of specifications,
                  which could result in patient skin burns or electrical
                  arcing.
                                     ----

Product:          Version V04 Software for Theraplan and TP-11 Treatment
                  Planning Systems, computerized workstations used to
                  calculate and determine appropriate radiation
                  treatment programs for doctors to administer cancer
                  patients.  Recall #Z-085-2.
Code:             Version V04.
Manufacturer:     Theratronics International Ltd., formerly Atomic Energy
                  of Canada, Ltd. (AECL Medical), Carrollton, Texas
                  (assembler of device/importer of components).
Recalled by:      Theratronics International, Ltd., formerly Atomic Energy
                  of Canada, Ltd. (AECL Medical), Kanata, Ontario, Canada,
                  by newsletter July 1987.  Firm-initiated field correction
                  ongoing.

                                      -8-



Distribution:     Nationwide and international.
Quantity:         131 units were distributed.
Reason:           The software version 4 was found to be deficient due
                  to many software "bugs".

                                     ----

Product:          Footrest for CGR Montery, TA50, TA60, Progrost, Bedford,
                  Delmar, Elstar, Ibis I and Ibis 2 Angulating X-ray Tables.
                  Recall #Z-094-2.
Code:             Undetermined.
Manufacturer:     GE Medical Systems, Inc., Waukesha, Wisconsin.
Recalled by:      Manufacturer, by letter October 3, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide.
Quantity:         690 units were distributed.
Reason:           The product was improperly manufactured without
                  spring pin holes and pins.  This could allow the
                  loaded footrest to slip off the table causing injury
                  to the patient and/or operator.

                                     ----

Product:          (a) National Medical Care Arterial Bloodlines, for use in
                  conjunction with a dialysis machines;
                  (b) National Medical Care On/Off diala-Kits, arterial
                  bloodline may be a component of the kit;
                  (c) Arterial Blood Tubing Sets, under Baxter label.
                  Recall #Z-097/099-2.
Code:             (a)  Catalog #03-9206-8   Lot numbers:  R1A010, R1A021,
                  R1A029, R1A063.  For use with the SPS/Baxter Dialysis unit.
                  1 injection site, post pump drip chamber, negative pressure
                  bloodline;
                  Catalog #03-9962-6  Lot #R1A097.  For universal use,
                  but high flow applications, pre and post pump drip chambers,
                  negative pressure bloodline;
                  Catalog #03-9228-2  Lot numbers:  R1A140, R0S157, R0S149.
                  For use with the Drake 480 and Fresenius 2008, with 6.5 mm
                  pump segment, Dialsyis Units.  Pre-pump drip chamber, 
                  negative pressure bloodline.
                  Catalog #03-9215-9  Lot numbers:  ROS166, ROS163.  For
                  universal use, post-pump drip chamber, negative pressure
                  bloodline;
                  Catalog #03-9201-9  Lot #R1B195 (addendum of 5/14/91),
                  ROS162, ROS148.  For universal use, pre-pump drip chamber,
                  negative pressure bloodline.
                  (b) Catalog #46-4370-6  Lot #M1B152;
                  Catalog #46-4662-6  Lot #M1B124;
                  (c) Catalog #96-4274-5  Lot #R51A650.  For unknown dialysis
                  machines, negative pressure, post-pump drip chamber.
                  Catalog #96-4285-1  Lot #R1A657.  For universal use,
                  negative pressure, post pump drip chamber;

                                      -9-



                  Catalog #96-4929-2  Lot #R1A660 (addendum of 5/14/91),
                  for universal use, negative pressure.
Manufacturer:     Erika of Texas, doing business as National Medical Care,
                  Medical Products Division, McAllen, Texas.
Recalled by:      National Medical Care, Medical Products Division,
                  Rockleigh, New Jersey, by letter dated April 20, 1991,
                  and May 15, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         64,896 units were distributed; firm estimates 32,000
                  units remain on the market.
Reason:           They may leak at the "T" connector.

                                     ----
                  Class III -

Product:          Becton Dickinson Microtainer brand Test Tubes (a)
                  containing K2 EDTA with mixing beads, reorder #5961;
                  (b) containing Ammonium Heparin, reorder #5963;
                  (c) containing K2 EDTA without mixing beads, reorder #5973,
                  in-vitro hematology/chemistry test reagent kits used to
                  collect, transport, anticoagulate and store skin puncture
                  blood specimens from infants, children, geriatric
                  patients and critical care patients.  Recall #Z-086/088-2.
Code:             (Consecutive manufactured lots beginning with lot number
                  prefixes: (a) OH, OJ, OK, OL, OM, 1A, 1B, 1C, 1D, 1E, 1F,
                  and 1G;
                  (b) 0B, 0C, 0D, 0E, 0F, 0G, 0H, 0J, 0K, 0L, 0M, 1A, 1B,
                  1C, 1D, 1E, 1F, 1G;
                  (c) OH, 0J, 0K, 0L, 0M, 1A, 1B, 1C, 1D, 1E, 1F, 1G.
Manufacturer:     Becton Dickinson Vacutainer Systems, San Lorenzo,
                  Puerto Rico.
Recalled by:      Becton Dickinson Vacutainer Systems, Rutherford,
                  New Jersey, by letter September 9, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide and international.
Quantity:         (a) 3,884,340 units; (b) 763,540 units; (c) 8,211,020
                  units were distributed.
Reason:           Use of the devices may lead to a delay in diagnosis due to
                  an inability to perform the tests.

                                     ----

Product:          3M Scotchcast 2 Casting Tape 2" width, product #72002
                  and 5" width, product #72005.  Recall #Z-093-2.
Code:             AUG91B29 on product boxes and AUG91B293 on pouches
                  containing individual rolls of tape.
Manufacturer:     3M Company, Irvine, California.
Recalled by:      3M Medical - Surgical Division, St. Paul, Minnesota,
                  by telephone September 30, 1991, followed by letter
                  October 18, 1991.  Firm-initiated recall complete.
Distribution:     Nationwide.
Quantity:         99 boxes (10 rolls per box) were distributed.

                                     -10-



Reason:           The product was produced with a urethane resin having
                  a silicone slip agent that was not intended for this
                  particular 3M product.

                                     ----

Product:          Becton Dickinson Unopette brand Test Kit containing
                  10 different concentrations of buffered saline, an
                  in-vitro Hemotology/Chemistry Test Reagent Kit for
                  Red Blood Cell (RBC), Osmotic Fragility determination
                  via manual methods.  Recall #Z-096-2.
Code:             Catalog number 5830, lot numbers: 0D444 EXP 10/15/91,
                  0J231 EXP 3/15/92, 0G132 EXP 1/15/92, 0H176 EXP 2/15/92,
                  0K273 EXP 4/15/92, 1B398 EXP 8/15/92, 1D442 EXP 10/15/92.
Manufacturer:     Becton Dickinson Vacutainer Systems, San Lorenzo,
                  Puerto Rico.
Recalled by:      Becton Dickinson Vacutainer Systems, Rutherford, New
                  Jersey, by letter September 9 and 10, 1991.  Firm-initiated
                  recall ongoing.
Distribution:     Nationwide and international.
Quantity:         35,830 units were distributed.
Reason:           Product may provide inaccurate test results.

                                     ----

                              VETERINARY PRODUCTS

                  Class I -
                                     NONE
                  Class II -

Product:          Methaplex Injection, B-Complex Injection Iso High Potency,
                  100 ml multi-dose product, used as a vitamin supplement.
                  Recall #V-007-2.
Code:             Lot #19257.
Manufacturer:     Walnut Pharmaceuticals, Inc., Anaheim, California.
Recalled by:      PRN Pharmacal, Inc., Pensacola, Florida, by letter
                  September 20, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide.
Quantity:         3,359 containers were distributed.
Reason:           Product was out of specification at the three month
                  stability check.  Product was subpotent in 
                  Vitamins B1, B2, and B12.

                                     ----

                  Class III -
                                     NONE

Injunction Actions Filed:

Against:          Adams Supply Company and Kathleen Jeffries,
                  Bedford, Indiana.

                                     -11-





Product:          Adams No-Mo Tape Worm Expeller for Dogs or any product
                  purporting to contain arecoline for animal use
                  (89-440-771).
Charge:           Misbranded - Product does not bear adequate directions
                  for use and cannot bear adequate directions, because
                  the drug is not safe for animal use except under the
                  professional supervision of a licensed veterinarian.
Filed:            August 15, 1991 - U.S. District Court for the Southern
                  District of Indiana, Indianapolis Division.
                  Civil # IP89768C, INJ #1212.

                                     ----
Seizure Actions Filed:

Product:          Retionic Acid Cream 0.05% and various topical drugs.
                  (91-571-145).
Charge:           Products are new drugs marketed without an approved
                  new drug application.
Firm:             C & M Pharmacal, Inc., Hazel Park, Michigan.
Filed:            August 21, 1991 - U.S. District Court for the Middle
                  District of Pennsylvania; Civil #ICV-91-1085,
                  FDC #66065.
                                     ----

Product:          Frozen Slipper Lobster Meat (91-640-461).
Charge:           Excessive glaze (ice) has been added to increase bulk
                  and weight; and the product is a food in package form
                  that fails to bear an accurate statement of quantity of
                  the content.
Firm:             Los Angeles Cold Storage, Los Angeles, California,
                  stored to Central Seaway, Inc., Northvield, Illinois.
Field:            October 30, 1991 - U.S. District Court for the
                  Central District of California; Civil #91-5924RSWL,
                  FDC #66296.
                                     ----