FDA Enforcement for the week of February 21, 1990 


                            FDA ENFORCEMENT REPORT
                             FOR FEBRUARY 21, 1990

                                                       February 21, 1990

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure to 
                  a violative product is not likely to cause adverse health 
                  consequences. 


                                      -1- 






                                   COSMETICS

                                      NONE

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 

                                      NONE

                  Class II -


Product:          Doxycycline Hyclate Capsules, 50 mg., in bottles of 50 and
                  500, an Rx antibiotic under the Bioline, Goldline and 
                  Superpharm labels.  Recall #D-147-0.
Code:             All lots. 
Manufacturer:     Superpharm Corporation, Bayshore, New York. 
Recalled by:      Manufacturer, by letter January 15, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     New York, California, Connecticut, Ohio, Florida, Texas.
Quantity:         26,036/50 capsule bottles were distributed. 
Reason:           Lack of assurance of bioequivalency and abbreviated New 
                  Drug Application discrepancies. 

                                      ----

Product:          (a) Recovered Plasma; (b) Recovered Plasma, Pooled. 
                  Recall #B-080/081-0.
Code:             Unit numbers: (a) C7006, C7007, C7010, C7013, C7014, C7018, 
                  C7021, C7022, C7025, C7029, C7032, C7033, C7035, C7037, C7039,
                  C7044, C7048, C7055, C7058, C7060, C7071, C7138, C8801, C8803,
                  C8874, C8875; (b) Pool numbers 3 (which includes Recovered
                  Plasma unit C8943); 4 (which includes Recovered Plasma, units 
                  C8946 and C8944); 8 (which includes Recovered Plasma, units 
                  C8949, C8948, C8756); 9 (which includes Recovered Plasma, 
                  units C8753, C8847, C8834; 10 (which includes Recovered 
                  Plasma, units C8759); 16 (which includes Recovered Plasma,
                  units C8796, C8806); 17 (which includes Recovered Plasma, 
                  units C8792 and C8797); 18 (which includes Recovered Plasma,
                  unit C8814); 20 (which includes Recovered Plasma units C8834, 
                  C8836, C8844, C8855, C8848); 22 (which includes Recovered 
                  Plasma, units C8843, C8850, C8853); 26 (which includes
                  Recovered Plasma unit C7049) and 28 (which includes Recovered 
                  Plasma units C7036, C7046, C7051, C7054). 
Manufacturer:     Community Hospital of Indianapolis, Indianapolis, Indiana.
Recalled by:      Manufacturer, by telephone December 7, 1989 followed by 
                  letter December 8, 1989.  Firm-initiated recall complete. 
Distribution:     Florida.

                                      -2- 





Quantity:         Firm estimates none remains on market.
Reason:           Blood components drawn from improperly screened donors
                  were distributed. 

                                      ----

Product:          Whole Blood.  Recall #B-082-0.
Code:             Unit #11G99752. 
Manufacturer:     American Red Cross, MO-Il Regional Red Cross Blood
                  Services, St. Louis, Missouri.
Recalled by:      Manufacturer, by telephone June 29, 1989.  Firm-initiated 
                  recall complete.
Distribution:     Missouri. 
Quantity:         1 unit was distributed and was destroyed. 
Reason:           Unit of Whole Blood was distributed that was drawn from a 
                  donor with a needlestick exposure 3 months prior to 
                  donation. 

                                      ----

Product:          (a) Red Blood Cells; (b) Fresh Frozen Plasma; 
                  (c) Platelets.  Recall #B-083/085-0.
Code:             Unit number J13498-0. 
Manufacturer:     St. Johns County Blood Bank, St. Augustine, Florida.
Recalled by:      Manufacturer, by letter October 30, 1989. 
                  Firm-initiated recall ongoing.
Distribution:     Florida.
Quantity:         1 unit of each component was distributed. 
Reason:           Blood products were distributed that were drawn from donors 
                  who previously tested repeatably reactive for the 
                  antibody to the human immunodeficiency virus (anti-HIV-1).

                                      ----

                  Class III - 

Product:          Rugby brand Triple Antibiotic Ointment in 1 ounce tubes,
                  OTC topical indicated for the use as a first aid to 
                  help prevent infection in minor cuts, scrapes and burns.
                  Recall #D-148-0.
Code:             Lot #910022.
Manufacturer:     Naska Pharmacal Co., Inc., Hauppauge, New York. 
Recalled by:      Manufacturer, by letter January 2, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         Firm estimates 1,327 to 1,376 tubes were involved in
                  the mix-up. 
Reason:           Some unit cartons are mislabeled as Bacitracin Ointment.

                                      ----

                                      -3- 





Product:          Nitroglycerin 2.6 mg Controlled Release Tablets, in 
                  bottles of 100 and in bulk (50,000 tablets per box),
                  an Rx vasodilator indicated for the prevention of 
                  angina pectoris due to coronary artery disease. 
                  Recall #D-149-0.
Code:             Lot numbers: IV8771W EXP 5/92 (bottles), 8771E EXP 2/89 
                  (bulk product). 
Manufacturer:     D.M. Graham Laboratories, Inc., Hobart, New York. 
Recalled by:      Rhone-Poulenc Pharmaceutical, Inc., Princeton, New
                  Jersey, by letter November 2, 1989 and November 
                  9, 1989.  Firm-initiated recall ongoing.
Distribution:     Nationwide, Belgium, The Netherlands. 
Quantity:         766/100 tablet bottles and 18/50,000 bulk containers
                  were distributed; firm estimates 100 bottles remain on
                  market. 
Reason:           Subpotency. 
                                      ----

Product:          Invamed brand Propranolol Hydrochloride Tablets, 20 mg.,
                  in 1,000 tablet bottles, an Rx oral cardiac depressant. 
                  Recall #D-150-0.
Code:             Lot #8M41 EXP 1/91. 
Manufacturer:     Invamed, Inc., Fairfield, New Jersey. 
Recalled by:      Manufacturer, by letter on or about January 31, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     New Jersey, California, Illinois. 
Quantity:         919,000 tablets were distributed; firm estimates 50 
                  bottles remain on market. 
Reason:           Product does not meet content uniformity requirements.

                                      ----

Product:          (a) Red Blood Cells; (b) Fresh Frozen Plasma. 
                  Recall #B-086/087-0.
Code:             Unit numbers J12513-5.
Manufacturer:     St. Johns County Blood Bank, St. Augustine, Florida.
Recalled by:      Manufacturer, by letter October 30, 1989. 
                  Firm-initiated recall ongoing.
Distribution:     Florida.
Quantity:         1 unit of each component was distributed. 
Reason:           Blood products were distributed that were drawn from
                  donors who previoviously tested repeatably reactive for 
                  the hepatitis B surface antigen (HbsAg).

                                      ----

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                      NONE

                                      -4- 





                  Class II -

Product:          General Purpose Lasers: (a) Model #48-1-115;
                  (b) Model #48-1-28; (c) Model #48-2-115.
                  Recall #Z-209/211-0.
Code:             Not coded.
Manufacturer:     Synrad, Inc., Bothell, Washington.
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan October 23, 1989.  Firm-initiated field 
                  correction ongoing. 
Distribution:     Nationwide. 
Quantity:         16 units were distributed.
Reason:           Noncompliance with the performance standard for laser 
                  products in that they lacked required remote interlock
                  connectors and manual reset mechanism.

                                      ----

Product:          Smart Monitor Models 900S, 950S, 970S, an Rx cardio-
                  respiratory monitor which, depending on the model, has a
                  memory capacity (900S has no memory capacity and 970S has 
                  the largest memory capacity).  Recall #Z-284/286-0. 
Code:             All lot numbers.
Manufacturer:     Healthdyne Technologies, Inc., Marietta, Georgia. 
Recalled by:      Manufacturer, by letter and field correction package
                  December 11, 1989.  Firm-initiated recall ongoing.
Distribution:     Nationwide. 
Quantity:         1,184 units were distributed. 
Reason:           Malfunctions in the power supply circuitry, caused by 
                  static electricity discharge, results in the monitors 
                  failing to function, and the internal alarm not sounding. 

                                      ----

                  Class III - 

                                      NONE

                              VETERINARY PRODUCTS 

                  Class I-

                                      NONE

                  Class II -- 

Product:          Various medicated feeds in 50 pound bags: 
                  (a) Medicated Pig Grower, "Alimentos El Agro," Tilosina 
                  40 g/t; 
                  (b) Medicated Broiler Starter, "Alimentos El Agro," Monensin
                  Sodium 100 g/t, bacitracin 25 g/t;
                  (c) Medicated Chick Starter, "Alimentos El Agro,: Monensin

                                      -5- 





                  Sodium 110 g/t; 
                  (d) Medicated Cock Fight, "Alimentos El Agro," Bacitracin 
                  50 g/t; 
                  (e) Starter/Growing (Chicken), "Alimentos El Agro," 
                  Bacitracin 10 g/t.  Recall #V-025/029-0.
Code:             All lots. 
Manufacturer:     Agro Ochoa, Inc., Guaynabo, Puerto Rico.
Recalled by:      Manufacturer, by telephone December 15, 1989.  Firm-
                  initiated recall ongoing. 
Distribution:     Puerto Rico.
Quantity:         (a) 2,434 bags; (b) 65 bags; (c) 95 bags; (d) 465 bags; 
                  (e) 537 bags were distributed.
Reason:           The products were not manufactured in conformance with
                  current good manufacturing practice, they may not contain 
                  the drug they are represented to contain, and they may be 
                  subpotent or superpotent. 

                                      ----
Seizure Actions Filed:

Product:          Latex Examination Gloves (90-552-584).
Charge:           Adulterated - Product contains holes; 
                  Mislabeled - Labeling fails to bear name and place of 
                  business of manufacturer, packer or distributor.
Firm:             Scitech Medical Products, Inc., New Hyde Park, New York.
Filed:            November 28, 1989 - U.S. District Court for the Western 
                  District of Missouri; Civil #89-1108-CV-W-3, FDC #65776.

                                      ----

Product:          Basmati Rice (89-566-970).
Charge:           Adulterated - Product contains insect, rodent and 
                  bird filth. 
Firm:             House of Spices, Buena Park, California.
Filed:            November 29, 1989 - U.S. District Court for the Central 
                  District of California; Civil #89-6894, FDC #65782. 

                                      ----

Product:          Chick Peas, Split Beans, Rice and Spices (90-576-444).
Charge:           Products were held under insanitary conditions. 
Firm:             Jai Hind Corporation, Chicago, Illinois.
Filed:            December 5, 1989 - U.S. District Court for the
                  Northern District of Illinois, Civil #89-C-8977,
                  FDC #65792. 
                                      ----

                                      -6- 





Product:          Foods in insect susceptible containers (90-526-663).
Charge:           Products were held under insanitary conditions. 
Firm:             Cauble and Field, Inc., Cape Girardeau, Missouri. 
Filed:            December 7, 1989 - U.S. District Court for the
                  Eastern District of Missouri; Civil #S89-0182-C,
                  FDC #65786. 

                                      ----

Product:          All foods in insect susceptible containers (90-566-488).
Charge:           Adulterated - Products contain insects. 
Firm:             PFC Foods, Inc., Los Angeles, California. 
Filed:            December 11, 1989 -- U.S. District Court for the Central
                  District of California; Civil #89-7136, FDC #65787. 

                                      ----

Product:          Latex Examination Gloves (89-455-605).
Charge:           Adulterated - Product contains holes; Misbranded -
                  Labeling fails to bear name and place of business of
                  manufacturer, packer, or distributor. 
Firm:             Voit Sports, Fairport, New York.
Filed:            December 14, 1989 - U.S. District Court for the District
                  of Nevada, Reno Division; Civil #CV-N-89-791-BRT, 
                  FDC #65789. 

                                      ----

Product:          All Foods in rodent susceptible containers (90-576-684).
Charge:           Products were held under insanitary conditions. 
Firm:             New Way Trading, Inc., Chicago, Illinois. 
Filed:            February 5, 1990 - U.S. District Court for the
                  Northern District of Illinois, Eastern Division;