Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure to 
                  a violative product is not likely to cause adverse health 
                  consequences. 


                                     NONE 
                                      -1- 






                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 

                  Class II -

Product:          (a) Platelets (expired); (b) Recovered Plasma.
                  Recall #B-127/128-0.
Code:             Unit numbers: (a) MC18462, MC15995; (b) MC13536, MC15736, 
                  MC18312, MC18437, MC18462, MC20667, MC21444, MC21602, 
                  MC21859, MC22326. 
Manufacturer:     Medical Center Hospital (MCH), Odessa, Texas. 
Recalled by:      Manufacturer, by letter January 16, 1990.  Firm-initiated 
                  recall complete.
Distribution:     (a) California; (b) Florida.
Quantity:         (a) 2 units; (b) 10 units were distributed. 
Reason:           Blood products were distributed which tested 1) 
                  repeatably reactive for HBsAg, 2) nonreactive for 
                  the antibody to the human immunodeficiency virus, 
                  type 1 (anti-HIV-1) by EIA but collected from donors
                  who had previously tested repeatably reactive for anti- 
                  HIV-1 by EIA. 

                                     ---- 

Product:          Cobe Spectra Tubing Set.
                  Recall #B-145-0.
Code:             Extended Life Platelet Tubing Set Catalog #777003-000, all
                  lots with an "S" in the lot number, and all lot numbers 
                  beginning with 01T, 02T, 03T, 04T.
Manufacturer:     Cobe Laboratories, Inc., Lakewood, Colorado.
Recalled by:      Manufacturer, by letter January 16, 1990.  Firm-initiated 
                  recall complete.
Distribution:     Nationwide and international. 
Quantity:         8,454 sets were distributed.
Reason:           Tubing sets were misassembled during manufacture which
                  resulted in the inversion of two lines connecting "y" 
                  connectors and the collection bags. 

                                     ---- 

                  Class III - 

Product:          Recovered Plasma.  Recall #B-129-0. 
Code:             Unit numbers: MC11720, MC15076, MC15084.
Manufacturer:     Medical Center Hospital, Odessa, Texas. 
Recalled by:      Manufacturer, by letter January 16, 1990.  Firm-initiated 
                  recall complete.
Distribution:     Florida, California.

                                      -2- 





Quantity:         3 units were distributed. 
Reason:           Recovered plasma was distributed which tested initially 
                  reactive for the hepatitis B surface antigen. 

                                      ----

Product:          Cobe Spectra Tubing Sets: (a) Platelet Set; 
                  (b) White Blood Cell Set.  Recall #B-146/147-0. 
Code:             (a) Catalog #777004-000, lot numbers: 08S15379, 
                  02T15378; (b) Catalog #777006-000, lot numbers: 
                  01T15394, 01T15388. 
Manufacturer:     Cobe Laboratories, Inc., Lakewood, Colorado.
Recalled by:      Manufacturer, by January 16, 1990.  Firm-initiated
                  recall complete.
Distribution:     Nationwide and international. 
Quantity:         (a) 456 sets; (b) 228 sets were distributed.
Reason:           Tubing sets were misassembled during manufacture which
                  resulted in the inversion of two lines connecting "y" 
                  connectors and the collection bags. 

                                      ----

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                      NONE

                  Class II -

Product:          Model 5800 CADD-PCA Ambulatory Infusion Pump, an Rx 
                  medical device intended for use as a pain management device.
                  Recall #Z-504-0.
Code:             Serial numbers 800501 through 804510. 
Manufacturer:     Pharmacia Deltec, Inc., St. Paul, Minnesota.
Recalled by:      Manufacturer, by letter January 24, 1990.  Firm-initiated 
                  field correction ongoing. 
Distribution:     Nationwide and Canada.
Quantity:         2,757 pumps were distributed. 
Reason:           Software error -- If the drug concentration entered into
                  the pump program is four mg/ml, the device will report
                  "MG Given" to be one-half the dose actually delivered.

                                      ----

Product:          Ballard Medical Products' Heimlich Micro Trach Transtracheal
                  Oxygen Delivery System, a single-use, sterile, 5F or 6F 
                  non-toxic catheter, intended to permit administration of
                  low flow oxygen to patients during surgery: 
                  (a) Catalog No. 170; (b) Catalog No. 171; 
                  (c) Catalog No. 175; (d) Catalog No. 176. 
                  Recall #Z-545/548-0.
Code:             Lot numbers: (a) 7 12 89 0021, D90-017, D90-022, D90-028; 
                  (b) D89-102, D90-016, D90-026, D90-032; 
                  (c) 7 12 89 0220, D90-018, D90-027; 
                  (d) D89-103, D90-014, D90-015, D90-021, D90-025, D90-033. 

                                      -3- 





Manufacturer:     Ballard Medical Products, Midvale, Utah.
Recalled by:      Manufacturer, by letter March 23, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Canada, Puerto Rico.
Quantity:         2,395 units were distributed. 
Reason:           The catheter may kink within the trachea. 

                                      ----

Product:          Medtronic Model 5866-9 Adaptor Kit containing an adaptor
                  (part 149039-G01) which is used with various non-Medtronic
                  leads to provide a compatible connection to a Medtronic 
                  Implantable Pulse Generator.  Recall #Z-552-0.
Code:             Lot #J89064064. 
Manufacturer:     Medtronic, Inc., Minneapolis, Minnesota.
Recalled by:      Manufacturer, by letter and facsimile March 16, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     Florida, Kentucky, Austria, Great Britain, The Netherlands, 
                  Spain, France, Italy, Portugal, West Germany. 
Quantity:         175 kits were distributed.
Reason:           An incorrect adaptor part that has a smaller inner
                  diameter in the female opening than the correct part
                  was placed in a portion of kits during packaging. 

                                      ----

Product:          (a) Models SCL-5130, SCL-5125, SCL 5025; (b) Model SPL-2120 
                  Laser Workstations, used for job shop application for 
                  cutting prototype parts, primarily automotive.
                  Recall #Z-611/612-0.
Code:             Serial numbers:  (a) HAY-0123 (SCL-5130), AAY-0106H 
                  (SCL-5125), HAY-0164H, HAY-0175H, HAY-0188H, HAY-0195H
                  (SCL-5025); (b) HAY-0151H.
Manufacturer:     Utilase Systems, Inc., Detroit, Michigan. 
Recalled by:      Manufacturer.  FDA approved the firm's corrective action
                  plan April 19, 1990.  Firm-initiated field correction 
                  ongoing.
Distribution:     Michigan, North Carolina, Illinois. 
Quantity:         8 units were distributed. 
Reason:           Noncompliance with the performance standard for laser 
                  products in that they lacked identification, certification
                  and other required labels.  Also, the user manual did not 
                  contain all the required information. 

                                      ----

Product:          Alpha MC-9 Series Mobile C-Arm Image Intensified Unit,
                  used for fluoroscopy with spot x-ray film capacity. 
                  Recall #Z-613-0.
Code:             Serial numbers:  MC7719, MC7721, MC7717, MC7711, MC7718,
                  MK7722, MC7724, MC8314. 
Manufacturer:     Instrumentarium Imaging, Inc., Milwaukee, Wisconsin.
Recalled by:      Manufacturer.  FDA approved the firm's corrective action
                  plan March 20, 1990.  Firm-initiated field correction 
                  ongoing.
Distribution:     Pennsylvania, Ohio, Kentucky, Delaware, Michigan. 
Quantity:         8 units were distributed. 

                                      -4- 





Reason:           Noncompliance with the Performance Standard for X-Ray 
                  Products 21 CFR 1020.31 (g)(l) "Field limitation and
                  alignment for spot-film devices."  The field size was 
                  not maintained when switching from fluoroscopy to 
                  radiography.

                                      ----

Product:          Alpha MC-9R C-Arm Image Intensified Unit, used for
                  job shop application for cutting prototype parts, 
                  primarily automotive.  Recall #Z-614-0. 
Code:             Serial number MC8314. 
Manufacturer:     Instrumentarium Imaging, Inc., Milwaukee, Wisconsin.
Recalled by:      Manufacturer.  FDA approved the firm's corrective action
                  plan March 30, 1990.  Firm-initiated field correction 
                  ongoing.
Distribution:     Michigan. 
Quantity:         1 unit was distributed. 
Reason:           Noncompliance with the performance standard for x-ray 
                  products.  The area of noncompliance include performance
                  specifications corresponding to mA and kV not meet the
                  accuracy statement at various kV settings.

                                      ----

Product:          Fisher Diagnostics Sera Chem Clinical Chemistry 
                  Control Serum (Bovine) Unassayed Level II, used as a
                  reference sample, intended to be used as if it were a 
                  clinical specimen (a positive control), intended for
                  estimating the precision and analytical bias in 
                  equipment used in the determination of a number of
                  clinical analysis.  Recall #Z-615-0.
Code:             Catalog #3111-84, Lot #N1280799 EXP 8/1/91. 
Manufacturer:     Instrumentation Laboratory, Division of Fisher
                  Scientific Company, Inc., Orangeburg, New York. 
Recalled by:      Manufacturer, by letter November 8, 1989.  Firm-initiated 
                  recall ongoing. 
Distribution:     Texas, Washington state, Tennessee, New Jersey, New York, 
                  Pennsylvania, California. 
Quantity:         1,027 packages were distributed.
Reason:           The product does not meet the label claim of 48 hours 
                  stability for Creatine Kinase (CK). 

                                      ----

Product:          3M PI-55 Urological 4.8 mm Surgical Staples, for use
                  in the creation of a continent ideal reservoir (Kock
                  Pouch) for urinary bladder replacement.  Recall #Z-616-0. 
Code:             JUL 89 104 89601. 
Manufacturer:     3M Edumex, Juarez, Mexico.
Recalled by:      3M Medical-Surgical Division, St. Paul, Minnesota, by 
                  telephone beginning April 4, 1990.  Firm-initiated
                  recall complete.

                                      -5- 





Distribution:     North Carolina, Illinois, Texas, California.
Quantity:         13 boxes of 12 individually-packaged staple cartridges
                  per box were distributed. 
Reason:           The staples were removed from the wrong end of the cartridge, 
                  which has differentiated ends because a hole for the
                  locking pin is in one end.

                                      ----

Product:          Medtronic Cardiac Sling, Model 7703PE, used during
                  cardiopulmonary bypass to hold the heart in an elevated 
                  position.  Recall #Z-626-0. 
Code:             Lot #PE89051140.
Manufacturer:     Wilshire Foam Products, Inc., Carson, California. 
Recalled by:      Medtronic, Inc., Fridley, Minnesota, by visit during
                  July 1989.  Firm-initiated recall complete. 
Distribution:     Wisconsin, Missouri.
Quantity:         20 units were distributed.
Reason:           The cardiac sling may shed fibers of polyurethane material
                  into the surgical wound due to improper die cutting of
                  the holes in the sling body.

                                      ----

Product:          Zenith-Cox Flexion Chiropractic Tables: 
                  (a) Model 90 (fixed height table);
                  (b) Model 95 (variable  height table).  Recall #Z-627/628-0.
Code:             Consecutive serial numbers 60454 through 62038, all units 
                  built between June 1988 and April 1989. 
Manufacturer:     Williams Manufacturing Company, Elgin, Illinois.
Recalled by:      Manufacturer, by letter April 17 and 23, 1990.
                  Firm-initiated field correction ongoing.
Distribution:     Nationwide and international. 
Quantity:         (a) 70 tables; (b) 298 tables were distributed; firm
                  estimates 90 percent remains on market. 
Reason:           One or more bolts in the caudal support linkage may 
                  work out during use, causing the caudal section of the
                  table to suddenly drop to the floor.

                                      ----

                  Class III - 

Product:          Dental Implant Impression Post, a component of the
                  Interpore IMZ Osteointegrated Implant System, used for
                  making dental fitting impressions.  Recall #Z-625-0.
Code:             Catalog #8425, lot #9281. 
Manufacturer:     Eberle Medizin, Technische Elemente GMBH, Wermberg, 
                  West Germany. 
Recalled by:      Interpore International, Irvine, California, by telephone 
                  November 8, 1989 followed by letter November 20, 1989.
                  Firm-initiated recall ongoing.
Distribution:     California, Florida, Ohio, Oregon.
Quantity:         17 posts were distributed.

                                      -6- 





Reason:           Size 3.3 x 2 mm posts were erroneously packaged and 
                  labeled as size 4.0 x 2 mm posts. 

                                      ----

Product:          Gas Cylinders for CDI Model 520 Calibrator, used to 
                  calibrate the CDI Model 200 and Model 300 Extracorporeal
                  Blood Gas Monitors: 
                  (a) CDI Gas I/Gas 1, Model No. 7210 Components
                  Concentration Contents 30.25 liters;
                  (b) CDI Gas II, Model No. 7240 Components Concentration 
                  Contents 30.25 liters.  Recall #Z-633/634-0.
Code:             Lot numbers:  (a) 013509, 033399, 053419, 080100, 120170, 
                  153629; (b) 010050, 020240, 053479, 060040, 083399. 
Manufacturer:     Chilton Metals, Chilton, Wisconsin. 
Recalled by:      CDI/3M Health Care, Irvine, California, by letter March 15, 
                  1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         1,573 cylinders were distributed. 
Reason:           The threads on the devices were not compatible with 
                  the threads on the CDI Model 520 Calibrator causing gas 
                  leakage which may result in miscalibration of the monitors. 

                                      ----

                              VETERINARY PRODUCTS 

                  Class I - 
                                      NONE

                  Class II -

Product:          MVI Lincomycin Premix for Swine and Chickens, 10 grams
                  per pound, in 50 pound bags.  Recall #V-043-0.
Code:             All lots. 
Manufacturer:     Mountaire Vitamins, Inc., North Little Rock, Arkansas.
Recalled by:      Manufacturer, by letter February 19, 1990.  Firm- 
                  initiated field correction complete.
Distribution:     Texas, Georgia. 
Quantity:         Firm estimated 20 bags remained on market.
Reason:           The product label failed to contain the required
                  warning and caution statements regarding the six-day
                  pre-slaughter withdrawal period, guaranteed analysis, 
                  and directions for use. 

                                      ----

                  Class III - 

Product:          Dexamethasone Tablets, USP, 0.25 mg, in 1,000 tablet
                  bottles, a veterinary Rx drug indicated for use as an 
                  anti-inflammatory agent in cats and dogs, under the 
                  following labels: Osborn Essar Corporation, Phoenix 
                  Pharmaceutical, Professional veterinary Labs, VEDCO,

                                      -7- 





                  Tech America, Central Vet Supply, The Butler Company. 
                  Recall #V-044-0.
Code:             Lot numbers: 206-138, 206-140, 206-141, 206-142,
                  206-143, 206-144, 206-146, 206-153. 
Manufacturer:     Sanofi Animal Health, Inc., La Seur, Minnesota. 
Recalled by:      Sanofi Animal Health, Inc., Ft. Dodge, Iowa, by letter
                  April 20, 1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide, Canada. 
Quantity:         12,300 bottles were distributed.
Reason:           Lots 146 and 153 are superpotent.  The rest are subpotent.

                                      ----

Medical Device Safety Alerts: 

Product:          Digit, Ear, and Flexible Sensors used with the LIFESTAT 
                  1600 pulse oximeter:
                  (a) Digit sensor, Part No. 804343-00; 
                  (b) Ear Sensor, Part No. 804582-00; 
                  (c) Flexible Sensors (adult, pediatric, infant and
                  neonate), Part numbers 804554-00, 804635-00,
                  804634-00, 804633-00.  Safety Alert #M-048/050-0. 
Code:             All sizes.  These assemblies are not serialized.
Manufacturer:     Physio Control Corporation, Redmond, Washington (digit
                  sensors); Honeywell Corporation, Juarez, Mexico (all
                  other sensors). 
Alerted by:       Physio Control Corporation, Redmond, Washington, by 
                  letter April 16, 1990.
Distribution:     Nationwide and international. 
Quantity:         5,362 (finger/hand); 1,817 (ear/hand); 6,285 (adult/flex);
                  4,523 (pediatric/flex); 1,107 (infant/flex); 2,998
                  (neonate/flex) sensors were distributed.
Reason:           When a Physio-Control sensor is used with Ohmeda pulse
                  oximeter models 3700, 3710, 3740, patient burns may occur 
                  at the sensor site.