FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
07/31/1990     

FDA Enforcement Report for the week of August 1, 1990 



                            FDA ENFORCEMENT REPORT
                              FOR AUGUST 1, 1990


                                                           August 1, 1990 

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.

                                     NONE 

                                      -1- 






                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 
                  Class II -

Product:          Source Plasma.  Recall #B-151-0.
Code:             Unit numbers:  221679, 483135, 625746, 625842, 626168.
Manufacturer:     Montgomery Plasma Center, Inc., Montgomery, Alabama.
Recalled by:      Manufacturer, by telephone on or about January 30, 1990 
                  and by letter January 31, 1990.  Firm-initiated recall
                  ongoing.
Distribution:     North Carolina. 
Quantity:         5 units were distributed. 
Reason:           Blood products which tested either 1) nonreactive for 
                  the antibody to the human immunodeficiency virus (anti- 
                  HIV-1) by EIA but collected from a donor who had
                  previously tested reactive for anti-HIV-1 by EIA or 
                  2) non-reactive for hepatitis B surface antigen (HBsAg) 
                  but collected from donors who had previously tested 
                  repeatably reactive for HBsAg, were distributed.

                                     ---- 

                  Class III - 
                                     NONE 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                     NONE 
                  Class II -

Product:          B-100 Argon Coagulator Disposable Footcontrol Pencil with 
                  10 ft. Cord, a single-use, sterile, electrosurgical Rx
                  device used in the operating room during patient
                  surgery.  Recall #Z-735/736-0.
Code:             All lots of catalog numbers 4009 and 4010.
Manufacturer:     Beacon Laboratories, Inc., Broomfield, Colorado.
Recalled by:      Manufacturer, by letter July 2, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         2,362 pencils were distributed; firm estimates 600 pencils
                  remain on market. 
Reason:           The ceramic tip may fall off. 

                                     ---- 

                                      -2- 





Product:          Endopath Disposable Surgical Trocar, a laparoscopic device
                  designed to fit 10/11 mm probes, used by the surgeon
                  during surgery.  Recall #Z-737-0. 
Code:             Catalog #T1011, lot numbers AP6882 and AP6884.
Manufacturer:     Ethicon, Inc. - Cincinnati, Ohio. 
Recalled by:      Ethicon, Inc., Somerville, New Jersey, by letter
                  February 27, 1990.  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         2,654 units were distributed; firm estimated 2,307 units
                  remained on market as of March 12, 1990.
Reason:           The gasket may separate or tear and dislodge or cause 
                  unintended desufflation.

                                    ----

Product:          Vernitron Majestic Table Top Sterilizers, Models 8080,
                  V8000, or R816 a table top steam autoclave with an
                  8" x 8" x 16" chamber.  Recall #Z-738/740-0.
Code:             All of the above model numbers manufactured prior to 1985.
Manufacturer:     Vernitron Better Built Corporation (OOB), Carlstadt,
                  New Jersey. 
Recalled by:      Vernitron Medical Products, also known as Ped-O-Jet 
                  International, Dayton, New Jersey, by letter February 
                  26, 1990.  Firm-initiated field correction ongoing. 
Distribution:     Nationwide, Canada, Korea.
Quantity:         Firm estimates fewer than 100 uncorrected units remain
                  on market.
Reason:           The locking hub may detach from the unit at high pressure 
                  as a projectile and may also cause a compromise of
                  sterility of the contents of the sterilizer.

                                    ----

Product:          Entube Enteral Feeding Tubes, non-sterile disposable
                  devices intended for single patient use to deliver
                  complete or supplemental liquid diets into the stomach or 
                  small bowel of patients having a functional G.I. tract: 
                  (a) Enteral Feeding Tube with Stylet; 
                  (b) Enteral Feeding Tube.  Recall #Z-742/743-0. 
Code:             Product No.     Lot No. 
                  (a) 121245      0002
                      921245      0001. 
Manufacturer:     Entech, Inc., Lebanon, New Jersey.
Recalled by:      Manufacturer, by letter March 26, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     California, Georgia, Illinois, Minnesota, New York, 
                  Texas, Missouri, Iowa, Wisconsin, North Dakota, Ohio. 
Quantity:         (a) 59 boxes (10 units/box); (b) 46 boxes (10 units/box)
                  were distributed. 
Reason:           The tungsten weighted tip can separate from the tube. 

                                    ----

                                    -3- 





Product:          AMS Dynaflex Penile Prosthesis, 15 mm diameter - all
                  lengths; part #72401148 (20 cm length) and part 
                  #72401149 (22 cm length).  Recall #Z-746-0. 
Code:             All lot numbers.
Manufacturer:     American Medical Systems Inc., Minnetonka, Minnesota. 
Recalled by:      Manufacturer, by voice mail notice May 11, 1990 and by
                  telephone during the period from May 11, 1990 to about
                  May 16, 1990.  Firm-initiated recall complete.
Distribution:     Alabama, Arkansas, California, Florida, Georgia,
                  Indiana, Michigan, North Carolina, Ohio, Tennessee, Texas,
                  Wisconsin, Washington, D.C. 
Quantity:         134 units or 67 pairs of the prostheses were distributed. 
Reason:           Device may leak hydraulic fluid after implantation. 

                                    ----

Product:          Sustain HA-Biointegrated Dental Implant System Fixed- 
                  Detachable Abutment Heads, 4.0/4.7 mm diameter: 
                  (a) Catalog #0440-4-05, 0.5 mm cuff length; 
                  (b) Catalog #0440-4-15, 1.5 mm cuff length. 
                  Recall #Z-749/750-0.
Code:             Lot Numbers:  (a) P0220-03; (b) P0220-04. 
Manufacturer:     Biointerfaces, Inc., San Diego, California. 
Recalled by:      Lifecore Biomedical/Orthomatrix, Inc., Minneapolis, 
                  Minnesota, by telephone June 14-15, 1990 followed by
                  letter June 28, 1990.  Firm-initiated recall ongoing. 
Distribution:     Maryland, New Mexico, The Netherlands.
Quantity:         (a) 3; (b) 6 abutment heads were distributed. 
Reason:           Potential for the units to fracture at the base of the
                  abutment heads. 

                                    ----

EXTENSION NOTE:   Carter Products' Trojan Naturalube Ribbed Condoms,
                  Recall #Z-092/094-0, which appeared in the November 22, 
                  1989 Enforcement Report has been extended to include
                  Trojan Naturalube Ribbed Condoms -  12's lot numbers
                  030989, X030989 and Trojan Naturalube Ribbed Condoms -
                  36's lot numbers 030989, 020889 and X020889.

                                    ----

                  Class III - 

Product:          Mon-A-Therm Airway Temperature Sensor.  Recall #Z-747-0.
Code:             Catalog #503-0640, lot numbers: 640LV50, 640AV50, 
                  640EV50, 640FV60, 640JV70, 640LV50, 640MV60, 640MV61, 
                  640A001, 640A002, 640A003, 640A004. 
Manufacturer:     Mon-A-Therm, Inc., New Athens, Illinois.
Recalled by:      Mallinckrodt Medical, Inc., St. Louis, Missouri, by 
                  letter July 2, 1990.  Firm-initiated recall ongoing.

                                    -4- 





Distribution:     Nationwide and international. 
Quantity:         703 cases were distributed. 
Reason:           Individual pouches were inadvertently labeled as sterile. 
                  The product is not intended to be used as a sterile 
                  device. 

                                    ----

Product:          DGR Intraocular Lens, Model No. PA17-OUV. 
                  Recall #Z-748-0.
Code:             Lot #A04011, serial numbers: 9112534/36, 9112575/77.
Manufacturer:     DGR, Inc., St. Petersburg, Florida. 
Recalled by:      Manufacturer, by telephone May 24, 1990.  Firm-initiated
                  recall complete.
Distribution:     Florida.
Quantity:         6 lenses were distributed; firm estimates none remains
                  on market.
Reason:           The lens power is mislabeled as 21.0 instead of 21.5. 

                                    ----

Product:          Precept Bouffant Caps, packaged 125/dispenser and 
                  4 dispensers per case.  Recall #Z-752-0.
Code:             Lot numbers 901001 - 936005 (all product distributed in 
                  1989).
Manufacturer:     Mars White Knight, Aqua Prieta, Somona, Mexico. 
Recalled by:      White Knight Healthcare, Asheville, North Carolina, by
                  letter September 25, 1989 followed by telephone.
                  Firm-initiated recall ongoing.
Distribution:     Nationwide, Canada, Saudi, Arabia.
Quantity:         Unknown.
Reason:           Skin irritation in wearers of the device due to the 
                  diammonium phosphate treatment necessary for these
                  devices to meet flammability requirements for wearing 
                  apparel.
                                    ----

                            VETERINARY PRODUCTS 

                  Class I - 

Product:          Bovi-Cu, Copper Disodium EDTA, in a 50:50 glycerine water 
                  mix, in 100 ml vials, a nonprescription veterinary
                  product developed for the prevention of copper deficiency 
                  in beef cattle and calves.  Recall #V-049-0.
Code:             All lots. 
Manufacturer:     Anthony Products Company, Arcadia, California.
Recalled by:      Veterinary Research and Development, Inc., Truckee, 
                  California, by letter dated June 15, 1990.  Firm-initiated
                  recall ongoing.  See also FDA press release P90-33, 
                  June 14, 1990.

                                    -5- 





Distribution:     California, Colorado, Idaho, Kansas, Missouri, Nebraska,
                  New Mexico, Oregon, Texas.
Quantity:         3,723 vials were distributed; firm estimates 2,211 vials
                  remain on market. 
Reason:           Copper toxicity in cattle following injection of
                  the product.

                                    ----
                  Class II -

                                    NONE

                  Class III - 

                                    NONE