FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
12/04/1990  

FDA Enforcement for the week of December 5, 1990



                            FDA ENFORCEMENT REPORT
                             FOR DECEMBER 5, 1990 

                                                            December 5, 1990

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.


Product:          Various brands of tuna salad: 
                  (a) Jack's Tuna Salad, 7 ounces, 12 pack; 
                  (b) Kinsers Tuna Salad, 12/6 pack;
                  (c) Kinsers Tuna Salad, 12 ounces, 24 pack; 
                  (d) Kinsers Tuna Salad, 24 ounces, 6 pack;
                  (e) Giles Tuna Salad, 2 pound container;
                  (f) Pocahontas Tuna Salad, 5 pound container; 
                  (g) Seaboard Homestyle Tuna, 5 pound, 2 pack; 
                  (h) Giles Homestyle Tuna Salad, 5 pounds, 2 pack; 
                  (i) Kinsers Tuna Salad, 5 pound container.
                  Recall #F-693-0.

                                      -1- 





Code:             Pull date 4/25/90.
Manufacturer:     Mrs. Giles Country Kitchen, Inc., Lynchburg, Virginia.
Recalled by:      Campbells Soup Company, Camden, New Jersey, by
                  telephone April 16, 1990.  Firm-initiated recall ongoing. 
Distribution:     Virginia, Kentucky, Florida, Alabama, Tennessee, Georgia, 
                  South Carolina, North Carolina. 
Quantity:         426 cases of various sizes and weights were distributed.
Reason:           Product is contaminated with Listeria monocytogenes.

                                     ---- 

Product:          (a) Old Fashioned Fruit Sherbet, Orange Flavored; 
                  (b) Lite Vanilla Flavored Ice Milk; 
                  (c) Chocolate Ice Cream.  Recall #F-721/723-0.
Code:             Lot numbers: (a) 902190; (b) 904040; (c) 906210.
Manufacturer:     J.J. Schmidt, Inc., Cincinnati, Ohio. 
Recalled by:      Sterling Ice Cream Company, Cincinnati, Ohio, by
                  telephone June 12, 1990 and by letter June 22, 1990.
                  Firm-initiated recall complete. 
Distribution:     Indiana, Kentucky, Ohio, Michigan.
Quantity:         Firm estimates none remains on the market.
Reason:           Product was contaminated with Listeria monocytogenes. 

                                     ---- 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.

                                     NONE 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                                     NONE 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 

Product:          Internal Defibrillation Handle Set, used with the 
                  Lifepak 9 Defibrillator/Monitor for internal
                  defibrillation.  Recall #Z-072-1. 

                                      -2- 





Code:             Date codes:  3-90 and 90-15.
Manufacturer:     Physio Control, Redmond, Washington.
Recalled by:      Manufacturer, by telephone beginning June 12, 1990
                  followed by letter June 13, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         142 units with code 3-90 and 106 units with code 90-15
                  were distributed. 
Reason:           An incorrect pin configuration in the cable that attaches 
                  the internal defibrillation hand set to the defibrillator 
                  adapter causes the Lifepak 9 unit to operate as if were 
                  configured for external defibrillation. 

                                     ---- 

                  Class II -

Product:          Radiological Therapy Planning (RTP) System
                  (a) Version 3A1; (b) Version 3B1.  Recall #Z-107/108-1. 
Code:             Serial numbers: 
                  RTP 5000     RTP 5000    RTP 5100     RTP 5200    RTP 5300
                  2104-3016A   2104-3106B   2104-3020A  2104-3021A  2104-3022A
                  112001       207016       311001      407010       408005 
                  112002       303029       406005      409008       503008 
                  202004       303030       409008
                  204015                    410009
                  211025                    411010

                  RTP 5400      RTP XL-23    RTP XL-23/73    RTP XL-73
                  2104-3023A   2104-3025    2104-3025D      2104-3026A

                  311006        702003       801005          511004 
                  311007                     805008          512006 
                  311008                     712002          512007 
                  311010                                     508001 
                  506002

                  RTP XL73     RTPXL73     RTP-XL73     RTP-XL172  RTP-XL 
                  2104-3026B  2104-3026C  2104-3026D   2104-3030A  2104-3032A 

                  603001      708001       709001       808001      808002
                  603002      708002       709002       809002
                  603003      708003       711003       809003
                  603005      708004       801006       810004
                  604006                   801007       810005
                  605007                   802007       901006
                  607008                   802008       901008
                  603008                   802009       904009
                  605010                   802010       906012
                  606011                   803011       906014
                  701013                   804012       907016
                  702015                   804013       906011
                  703016                   712005 

                                      -3- 





                  RTP-XL
                  2104-3033A

                  807001
                  807002
                  906004. 
Manufacturer:     ADAC Laboratories, Milpitas, California.
Recalled by:      Manufacturer, by letter October 30, 1989.  Firm-initiated 
                  field correction complete.
Distribution:     Nationwide, Thailand. 
Quantity:         76 systems were distributed.
Reason:           The software program contains a calculation error which 
                  may not provide user with all required information. 

                                     ---- 

Product:          PolySafe Implantable Endocardial Pacing Leads 
                  (a) Model BV-317, Ventricular biopolar, trabeculok
                  fixation connector; (b) Model SV-309, Ventricular unipolar
                  trabeculok fixation connector.  Recall #Z-109/110-1.
Code:             (a) All serial numbers; (b) All leads with serial 
                  numbers prefaced by 309L. 
Manufacturer:     Cardiac Control Systems, Palm Coast, Florida. 
Recalled by:      Manufacturer, by telephone and letter September 19, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     (a) Texas, New Jersey, Oklahoma, Florida, Maine,
                  Louisiana, New York, Michigan, Pennsylvania, Connecticut; 
                  (b) Texas, New Jersey, New York, Florida, Louisiana,
                  Arkansas, Massachusetts.
Quantity:         (a) 74 units; (b) 33 units were distributed.
Reason:           The lead connector that attaches the proximal end of the
                  lead to the pulse generator unit is too big in diameter.
                  Excess adhesive was used to affix the connector to the
                  silicon seal between the sheath and connector.

                                     ---- 

CORRECTION:       Lastac System, Model 1025 Laser Product, Recall #Z-022-1
                  which appeared in the November 7, 1990 Enforcement
                  Report should read: 

                  Reason:  Noncompliance with performance standard for laser
                  products in that it lacks a Class IV warning logo-type
                  label adjacent to the laser output aperture on the
                  Model 1025 Control Monitor. 

                                     ---- 

                                      -4- 





                  Class III - 

Product:          Coulter JT2 and Coulter JT3 Automated Hematology
                  Analyzer, instruments needed in clinical and hospital 
                  labs for hematology studies on human blood. 
                  Recall #Z-101/102-1.
Code:             Catalog numbers: 664073, 6604123, 6604124, 6604125, 
                  6604142, 6604143, 6604144, 6604145.  Part numbers:
                  6704765, 6704816, 6701817, 6704820, 6704874, 6704875, 
                  6704876, 6704877.  Serial numbers: 0 through S35590.
Manufacturer:     Coulter Electronics, Inc., Hialeah, Florida.
Recalled by:      Manufacturer, by letter September 7, 1990.  Firm- 
                  initiated field correction ongoing. 
Distribution:     Nationwide and international. 
Quantity:         (a) 166 units; (b) 248 units were distributed.
Reason:           The device has the potential to dilute patient samples
                  and also to loose pre-set calibration factors under 
                  certain conditions such as power interruption.

                                     ---- 

                              VETERINARY PRODUCTS 

                  Class I - 
                                     NONE 

                  Class II -

Product:          Phoenix Pharmaceutical brand Mineral Oil Light USP, in
                  1 gallon brown plastic containers, a pharmaceutic aid for 
                  veterinary use only.  Recall #V-050-1.
Code:             Lot #590. 
Manufacturer:     Tab Chemical, Chicago, Illinois.
Recalled by:      Hydrox Chemical Company, Elk Grove Village, Illinois; 
                  by telephone October 29, 1990.  Firm-initiated recall 
                  ongoing.
Distribution:     Florida, Maryland, Pennsylvania.
Quantity:         56 cases were distributed; FDA estimates 25 percent 
                  remains on market.
Reason:           Some of the oil is contaminated with propylene glycol.

                                     ---- 

                  Class III - 
                                     NONE 

Injunction Action Filed:

Against:          Richards Bros., Inc., Mountain Grove, Missouri. 
Product:          Multiple Foods (90-553-185/193).
Charge:           Adulterated -- Products were held for sale in a warehouse 
                  infested with insects and rodents.

                                      -5- 





Filed:            November 27, 1990 - U.S. District Court for the Western 
                  District of Missouri, Southern Division;
                  Civil #90-3514-CV-S-4, INJ #1244. 

                                     ---- 

Seizure Actions Filed:

Product:          Various Articles of Food (91-575-474).
Charge:           Adulterated -- Products were held for sale in a warehouse 
                  infested with insects and rodents.
Firm:             Indian Groceries and Spices, Inc., Skokie, Illinois.
Filed:            November 26, 1990 - U.S. District Court for the 
                  Northern District of Illinois, Eastern Division;
                  Civil #90C6819, FDC #65974. 

                                     ---- 

Product:          Dessicated Coconut (90-609-718).
Charge:           Adulterated - Product was held under insanitary 
                  conditions whereby it may have become contaminated
                  with filth. 
Firm:             Virmurco, Medina, Ohio. 
Filed:            November 26, 1990 - U.S. District Court for the 
                  Northern District of Ohio, Eastern Division;
                  Civil #5:90CV2098, FDC #65961.