FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
12/18/1990 

FDA Enforcement for the week of December 19, 1990 



                            FDA ENFORCEMENT REPORT
                            FOR DECEMBER 19, 1990 

                                                    December 19, 1990 
Recalls and Field Corrections:

                                    FOODS 

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.


Product:          Pride of the Shore Crabmeat in 16 ounce containers: 
                  (a) Claw Crabmeat; (b) Special Crabmeat; (c) Lump 
                  Crabmeat; (d) Backfin Crabmeat.  Recall #F-673/676-0. 
Code:             Not coded.
Manufacturer:     Eastern Shore Seafood Company, Inc., Onancock, Virginia.
Recalled by:      Manufacturer, by telephone July 18, 1990.  Firm-initiated 
                  recall complete.
Distribution:     Maryland, Virginia. 
Quantity:         Firm estimates none remains on the market.
Reason:           Products were contaminated with Listeria monocytogenes. 

                                     ---- 

                                     -1-






Product:          Pride of the Shore Crabmeat in 16 ounce containers: 
                  (a) Claw Crabmeat; (b) Special Crabmeat; (c) Lump 
                  Crabmeat.  Recall #F-011/013-1. 
Code:             Not coded.  Product processed 8/22/90.
Manufacturer:     Eastern Shore Seafood Company, Inc., Onancock, Virginia.
Recalled by:      Manufacturer, by telephone September 17, 1990.
                  Firm-initiated recall complete. 
Distribution:     Virginia, Pennsylvania. 
Quantity:         Firm estimates none remains on the market.
Reason:           Products were contaminated with Listeria monocytogenes. 

                                     ---- 

Product:          Keyser Brothers brand (a) Fresh Claw Crabmeat, in 
                  5 pound plastic bags; (b) Frozen Claw Crabmeat, in
                  5 pound plastic bags.  Recall #F-014/015-1. 
Code:             Some bags are uncoded and some are coded 9/25/90. 
Manufacturer:     Keyser Brothers, Inc., Lottsburg, Virginia. 
Recalled by:      Manufacturer, by telephone September 26, 1990 followed
                  by letter.  Firm-initiated recall ongoing.
Distribution:     Maryland, Virginia, Ohio, Pennsylvania, South Carolina. 
Quantity:         Approximately 1,000 pounds were distributed.
Reason:           Products are contaminated with Listeria monocytogenes.

                                     ---- 

Product:          (a) Ham 'N' Cheese Sandwich, 3.75 ounces, packaged in a 
                  polypropylene overwrap with yellow stick-on label;
                  (b) Ham and Swiss Cheese Croissant, 3 3/4 ounces. 
                  Recall #F-034/035-1.
Code:             (a) 2762; (b) 081.
Manufacturer:     Stewart Sandwiches, Inc., Norfolk, Virginia.
Recalled by:      Manufacturer, (a) by telephone October 26, 1990;
                  (b) by telephone June 27, 1990.  Firm-initiated recall
                  ongoing.
Distribution:     (a) Southeastern United States; (b) Maryland. 
Quantity:         (a) 11,232 sandwiches (b) 1,980 croissants
                  were distributed; firm estimates none remains on
                  the market. 
Reason:           Products are contaminated with Listeria monocytogenes.

                                     ---- 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                                      -2- 





                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          Flowing Wells Pure Drinking Water, in 16 ounce and
                  in 1 gallon plastic containers.  Recall #F-016-1. 
Code:             Lot numbers: 051891, 062591, 062691, 091990, 062690.
Manufacturer:     Southern Beverage Packers, Agusta, Georgia. 
Recalled by:      Manufacturer, by visit October 10, 1990.
                  Firm-initiated recall ongoing.
Distribution:     Georgia, South Carolina, North Carolina.
Quantity:         20,000 gallon bottles and 817 cases of 16 ounce containers
                  were distributed. 
Reason:           Product is contaminated with mold.

                                     ---- 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 
                  Class II -

                                     NONE 
                  Class III - 

EXTENSION NOTE:   Conjugated Estrogen Tablets, 0.625 mg, manufactured 
                  by Duramed Pharmaceuticals, Inc., Recall #D-089-1 
                  which appeared in the November 7, 1990 Enforcement
                  Report has been extended to include the following lot 
                  numbers:  91696, 91707, 91705, 91703, 91695, 92024, 
                  91698, 92304, 92332, 92333, 92337, 92339, 92335, 92349, 
                  92346, 91714, 91716, 91694, 91697, 92338. 

                                     ---- 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 
                                     NONE 

                  Class II -

Product:          Parallel Hole Collimator for Gamma Camera, Siemens
                  Model ZLC 0.070. 10 mil., 1.0".  Recall #Z-083-1. 
Code:             No lot or serial numbers. 
Manufacturer:     Nuclear Technologies, Inc., New Haven, Connecticut. 

                                      -3- 





Recalled by:      Manufacturer, by telephone August 27, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Georgia, Ohio.
Quantity:         4 units were distributed. 
Reason:           Threaded inserts failed in collimator housing causing 
                  collimator to fall off the camera.

                                     ---- 

Product:          (a) Pediatric Urine Meter without Catheter; 
                  (b) Pediatric Urine Meter with 5 FR Catheter; 
                  (c) Pediatric Urine Meter with 8 FR Catheter. 
                  Recall #Z-093/095-1.
Code:             Catalog No.      Lot No.
                  (a) 153206       76J9X275, 76DA2330 
                  (b) 895605       76D9X246A, 76H9X573
                  (c) 895608       76J9X277, 76J9X278, 76AA2275, 76DA2329.
Manufacturer:     Bard Urological Division, Sonora, Mexico. 
Recalled by:      C.R. Bard Urological Division, Covington, Georgia, by 
                  telephone September 6, 1990 followed by letter. 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide, Canada. 
Quantity:         1,255 cases were distributed. 
Reason:           The product may have blocked drainage tubes which prevents
                  proper drainage of the bladder. 

                                     ---- 

Product:          Cobe Precision Pump Rear Panel Breaker/Harness Assembly,
                  a component of the Cobe Precision Blood Pump, which is a
                  microprocessor controlled device used to control and
                  monitor blood flow rates during surgery.  Recall #Z-100-1.
Code:             Catalog #043600-000, serial numbers:
                  3105AP     3120AP     3141AR     3151AR 
                  3106AP     3129AR     3143AP     3152AR 
                  3107AP     3130AR     3144AR     3153AR 
                  3108AP     3131AP     3145AR     3154AR 
                  3110AP     3132AR     3146AR     3155AR 
                  3113AP     3133AR     3147AR     3156AR 
                  3114AP     3138AR     3148AR
                  3116AP     3139AR     3149AR
                  3117AP     3140AR     3150AR
Manufacturer:     COBE Laboratories, Inc. Cardiovascular Division,
                  Arvada, Colorado. 
Recalled by:      Manufacturer, by letter October 1, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Utah, California, Illinois, Texas, Georgia, Pennsylvania. 
Quantity:         33 pumps were distributed.
Reason:           Circuit breaker on the back panel pump is improperly wired
                  resulting in only one winding of the primary transformer
                  being connected to main voltage.
                                     ---- 

                                      -4- 





Product:          Hirider Powered Wheelchairs.  Recall #Z-112-1.
Code:             All serial numbers. 
Manufacturer:     Retec P.R. Inc., Guayama, Puerto Rico.
Recalled by:      Retec USA Inc., Orchard Park, New York, by letter 
                  October 4, 1990.  Firm-initiated field correction ongoing.
Distribution:     Nationwide, Canada, England.
Quantity:         174 units were distributed. 
Reason:           The wheelchairs are subject to uncommanded movements
                  due to possible malfunction in the joystick module. 

                                     ---- 

Product:          IRIS Oculight SL Laser Diode Endophotocoagulator. 
                  Recall #Z-113-1.
Code:             Serial numbers: Undetermined. 
Manufacturer:     Iris Medical Instruments, Inc., Mountain View,
                  California. 
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan November 6, 1990.  Firm-initiated field 
                  correction ongoing. 
Distribution:     Nationwide. 
Quantity:         22 units were distributed.
Reason:           Device is noncompliant with the Radiation Control for 
                  Health and Safety Act due to the lack of an interlock 
                  at the fiberoptic connector, the lack of a calibration
                  procedure in the user manual and inappropriate placement of 
                  the noninterlocked protective housing label on the product
                  housing.
                                     ---- 

Product:          R-1000 Automated Reticulocyte Analyzer, a medical laser 
                  product.  Recall #Z-114-1.
Code:             Serial numbers: Undetermined. 
Manufacturer:     TOA Medical Electronics, Kobe, Japan. 
Recalled by:      TOA Medical Electronics (USA), Inc., Los Alamitos,
                  California.  FDA approved the firm's corrective action
                  plan November 25, 1990.  Firm-initiated field correction
                  ongoing.
Distribution:     Nationwide. 
Quantity:         31 units were distributed.
Reason:           The device is noncompliant with the Radiation for 
                  Health and Safety Act in that the units allowed human 
                  access to Class IIIb level of laser light.  Also, the 
                  products did not have the correct text on the 
                  noninterlocked protective housing labels, and the user
                  information did not have the appropriate safety information 
                  and label reproduction. 

                                     ---- 

                                      -5- 





Product:          AMK Total Knee System with Metal Backed Patella 
                  Component, 9.5 mm thick by 38 mm diameter, a sterile
                  device intended for implant.  Recall #Z-118-1.
Code:             Product #1488-18-000, lot #668072***, where *** is a
                  sequential number between 001 through 012.
Manufacturer:     DePuy, Inc., Division of Boehringer-Mannheim Corporation, 
                  Warsaw, Indiana.
Recalled by:      Manufacturer, by telephone September 10, 1990.
                  Firm-initiated recall ongoing.
Distribution:     Nationwide. 
Quantity:         12 units were distributed.
Reason:           The device was labeled as product 1488-18 on one end
                  and product 1488-17 on the other end.  The two products 
                  differ in size. 

                                     ---- 

Product:          Fehling Disposable Bioptome, an endomyocardial
                  instrument.  The recall covers all product licensed to DLP, 
                  Inc. for packaging, sterilizing and distribution.  This 
                  includes 10 different model numbers differentiated by shaft 
                  length and cup outside diameter:
                      Model Number                  Shaft        Cup O.D. 
                  (a) 41116  Disposable Bioptome    510 mm       1.6 MM 
                  (b) 41118  Disposable Bioptome    510 mm       1.8 MM 
                  (c) 41122  Disposable Bioptome    510 mm       2.2 MM 
                  (d) 41216  Disposable Bioptome    800 mm       1.6 MM 
                  (e) 41218  Disposable Bioptome    800 mm       1.8 MM 
                  (f) 41316  Disposable Bioptome    1000 mm      1.6 MM 
                  (g) 41318  Disposable Bioptome    1000 mm      1.8 MM 
                  (h) 41416  Disposable Bioptome    1200 mm      1.6 MM 
                  (i) 41418  Disposable Bioptome    12000 mm     1.8 MM 
                  (j) 41422  Disposable Bioptome    12000 mm     2.2 MM 
                  Recall #Z-121/130-1.
Code:             All model numbers and all lot numbers:
                  Lot numbers:  (a) 6890-5002, 6890-7005, 6900-3002, 6900-6001
                  (b) 6900-1009, 6900-3002
                  (c) 6890-7008, 6900-3002, 6900-6002, 6900-8001
                  (d) 6900-3002 
                  (e) 6900-3005, 6900-8001
                  (f) 6900-3002 
                  (g) 6900-7001 
                  (h) 6890-7008, 6900-3002, 6900-5009 
                  (i) 6900-3002, 6900-8001
Manufacturer:     Fehling Medizintechnik GMBh, Karlstein, West Germany. 
Recalled by:      DLP, Inc., Grand Rapids, Michigan, by telephone 
                  October 3, 1990 followed by letter October 5, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide. 
Quantity:         319 units were distributed; firm estimates 50 units 
                  remain on the market. 

                                      -6- 





Reason:           The jaws of the device will not close after multiple uses 
                  in the same patient because the core wire becomes 
                  disengaged. 
                                     ---- 

Product:          Insall/Burstein II Modular Knee System, product #5220-31-03,
                  posterior stabilized series, tibial articular surface, a
                  component of a knee prosthesis.  Recall #Z-151-1. 
Code:             Lot #69761700.
Manufacturer:     Zimmer, Inc., Warsaw, Indiana.
Recalled by:      Manufacturer, by letter October 30, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, France, Japan, England, Spain.
Quantity:         59 pieces subject to recall.
Reason:           The word "BACK" was incorrectly stamped onto the front
                  surface of the device.

                                     ---- 

CORRECTION:       Respironics BIPAP Ventilatory Support Systems, Airway 
                  Management Systems, and Remote Control Systems, 
                  Recall #Z-054/061-1 which appeared in the November 14,
                  1990 Enforcement Report should read:

                  Code:  Various serial numbers ranging from 3801 through 
                         and including 5850.

                                     ---- 

                  Class III - 

Product:          Sterile Disposable Microhone Needles, Bunnels Straight, 
                  Model 220002.  Recall #Z-104-1. 
Code:             90D06.
Manufacturer:     Richard-Allen Medical Industries, Inc., Richland, Michigan. 
Recalled by:      Manufacturer, by telephone on or about July 11, 1990. 
                  Firm-initiated recall complete. 
Distribution:     Michigan, Wisconsin, California, Colorado.
Quantity:         864 needles were distributed. 
Reason:           The package labeled as Bunnels Straight, Model 220002 
                  but actually contained Model 2222 curved needles. 

                                     ---- 

Product:          PRO SOL CHX (Chlorhexadine) Antimicrobial Irrigating
                  Medicament, in 16 ounce bottles, formulated for use with
                  Cavi-Med System, an Rx product used during dental procedures. 
                  Recall #Z-115-1.
Code:             051490. 
Manufacturer:     L.D. Caulk Company, Milford, Delaware.
Recalled by:      Dentsply International Inc., York, Pennsylvania, by 
                  letter October 3-4, 1990.  Firm-initiated recall ongoing. 
Distribution:     Nationwide. 

                                      -7- 





Quantity:         2,532 bottles were distributed; firm estimates 442 bottles
                  remain on the market. 
Reason:           Product contains cinnamon oil in the ingredient alcohol 
                  instead of peppermint oil as stated on the label. 

                                     ---- 

                              VETERINARY PRODUCTS 

                  Class I - 
                                     NONE 

                  Class II -

Product:          McNess Bulk Custom Dairy Supplement, in 2,000 pound 
                  bags.  Recall #V-051-1. 
Code:             Control #0J890. 
Manufacturer:     Furst-McNess Company, Freeport, Illinois. 
Recalled by:      Manufacturer, by telephone October 15, 1990.
                  Firm-initiated recall ongoing.
Distribution:     Illinois. 
Quantity:         firm estimates 2 bags remain on the market. 
Reason:           Selenium content is superpotent.

                                     ---- 

                  Class III - 

Clarification Note: 

                  Anti-Globulin Control (AGS) Serum, manufactured by
                  Immuno-Mycologics, Recall #Z-064-1 which appeared in the
                  November 14, 1990 Enforcement Report should read: 

                  Reason:  As a result of a failure in the Anti-Globulin
                           Control Lot 10IR2, it does not become negative 
                           after the detacher enzyme treatment, thus
                           making the detacher enzyme appear defective
                           erroneously. 
                                     ---- 

Medical Device Safety Alerts: 

Product:          Omni-Flow I.V. Pump, Model No. 4000, designed to deliver
                  up to four different medications either continuously or 
                  intermittently.  Safety Alert #M-005-1. 
Code:             All serial numbers. 
Manufacturer:     Omni-Flow, Inc., Wilmington, Massachusetts. 
Alerted by:       Abbott Laboratories, Hospital Products Division, Abbott 
                  Park, Illinois, by letter November 6, 1990. 
Distribution:     Nationwide, Canada. 
Quantity:         Approximately 5,000 units were distributed. 

                                      -8- 





Reason:           The infusion pump generates electronic noise, i.e. static 
                  electricity, during operation which may create artifacts
                  on patient monitoring devices, such as ECG, EMG and EEG 
                  machines.  These artifacts, if not recognized as such, may
                  lead to an erroneous interpretation of the readings from
                  the patient monitoring. 

                                     ---- 
Seizure Actions Filed:

Product:          Enriched Rice (90-638-721). 
Charge:           Adulterated - Product is contaminated with rodent urine.
                  In addition product was held in a rodent infested warehouse 
                  in bags bearing rodent excreta pellets and rodent urine 
                  stains. 
Firm:             Havanera Tropical Market Corp., Bronx, New York.
Filed:            June 29, 1990 - U.S. District Court for the Southern
                  District of New York; Civil #90 CIV 4342 (PRL), FDC #65873. 

                                     ---- 

Product:          Curry Powder and Other Foods (90-626-628/90-626-661). 
Charge:           Products were held under insanitary conditions whereby
                  they may have become contaminated with filth. 
Firm:             Glendale Warehouse, Brooklyn, New York/ 
                  Manhattan Milling & Drying Co., Brooklyn, New York. 
Filed:            October 1, 1990 - U.S. District Court for the Eastern 
                  District of New York; Civil #CV-90-3391/CV-90-3392, 
                  FDC #65929/65935
                                     ---- 

Product:          Canned Fish (89-353-331). 
Charge:           Product was prepared, packed or held under insanitary 
                  conditions whereby it may have been rendered
                  injurious to health.
Firm:             Gold Star Smoked Fish, Brooklyn, New York.
Filed:            April 25, 1990 - U.S. District Court for the Eastern
                  District of New York; Civil #CV-90-1404, FDC #65841.

                                     ---- 

Product:          Sesame Seeds (90-638-673).
Charge:           Adulterated - Product was stored in a warehouse with
                  extensive bird activity.
Firm:             Aceste Trucking Company, Brooklyn, New York.
Filed:            September 7, 1990 - U.S. District Court for the 
                  Eastern District of New York; 
                  Civil #CV-90-3104, FDC #65907.

                                     ---- 

                                      -9- 





Product:          Anaform Femoral Endoprothesis and Tempromandibular Joints 
                  (90-451-984/5). 
Charge:           Anaform (hip) implant -- Adulterated under 501(f)(1)(B) 
                  in that it is a Class III device distributed without an 
                  approved PMA aand is not exempt from such approval. 
                  Temporomadibular Joints (jaw) implants are misbranded 
                  under 502(j), 502(f)(1) and 502(o) in that they are 
                  dangerous to health when used in the manner or duration 
                  prescribed, recommended, or suggested in the labeling; its
                  labeling fails to bear adequate directions for use and
                  failure to file notice under 510(k);
Firm:             Novamed, Inc. and Oral Surgery Marketing Inc.,
                  Houston, Texas. 
Filed:            September 20, 1990 - U.S. District Court for the Southern 
                  District of Texas, Houston Division; Civil #CA H-90-3000, 
                  FDC #65918. 
                                     ---- 

Product:          Anaform Femoral Endoprothesis (90-451-986). 
Charge:           Adulterated within 501(f)(1) because they are Class III 
                  devices distributed without premarket approval and they 
                  are not exempt from such approval.
Firm:             Novamed, Inc., Houston, Texas.
Filed:            September 20, 1990 - U.S. District Court for the
                  Southern District of Texas, Houston Division; 
                  Civil #CA H-90-3001, FDC #65919.

                                     ---- 

Product:          Temporomandibular Joints (90-604-175).
Charge:           TMJs are misbranded under 502(j), 502(f)(1) and 502(o)
                  in that they are dangerous to health when used in the manner
                  or duration prescribed, recommended, or suggested in the
                  labeling; its labeling fails to bear adequate directions
                  for use and failure to file notice under 510(k);
Firm:             Vitek, Inc., Houston, Texas.
Filed:            September 20, 1990 - U.S. District For the Northern 
                  District of Texas, Dallas Division; Civil #CA 3-90-2142-G,
                  FDC #65920.