The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 ENFORCE
02/19/1991 

FDA Enforcement for the week of February 20, 1991.



                            FDA ENFORCEMENT REPORT
                               FOR FEB. 20, 1991
                                                         February 20, 1991
Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 


Product:          Phoenix Cookies:
                  (a) Chocolate Chip Cookie;
                  (b) Oatmeal Raisin Cookie;
                  (c) Vanilla Cookie. Recall #F-688/690-0.
Code:             All lots. 
Manufacturer:     Phoenix International Marketing Corporation, Sparks, Nevada.
Recalled by:      Manufacturer, by letter May 15, 1990.  Firm-initiated recall
                  ongoing.

                                      -1- 





Distribution:     Nationwide. 
Quantity:         1,311,559 cases were distributed. 
Reason:           Products made unsubstantiated drug claims and their 
                  sugar content is well in excess of label claim. 

                                     ---- 

Product:           Our Family Mushrooms, Pieces and Stems, in 4 ounce 
                   cans.  Recall #F-132-1.
Code:              Lot numbers: MBSP/SF/0621, MBSP/SF/0623, MBSP/SF/0624. 
Manufacturer:      Sam Yang Moolsan Company, Ltd, Republic of Korea 
Recalled by:       Worldwide Foods, Bloomington, Minnesota, by letter 
                   November 29, 1990.  Firm-initiated recall ongoing. 
Distribution:      North Dakota, Minnesota. 
Quantity:          6,000 cases (24/4 ounce cans/case)  were distributed.
Reason:            Possible contamination with Staphylococcus aureus organisms. 

                                     ---- 

Product:          Pacific Friend Canned Mushrooms, Stems and Pieces,
                  in 4 ounce cans.  Recall #F-133-1.
Code:             Lot numbers 0618 and 0620.
Manufacturer:     Mulsen Trading Company, Ltd., Seattle Washington
Recalled by:      Manufacturer, by telephone and by letter December 27, 1990. 
                  Firm-initiated recall ongoing.
Distribution:     Oregon, Washington, Idaho, Akansas, California, Montana.
Quantity:         9,200 cases (24 cans/case) were distributed.
Reason:           Possible contamination with Staphylococcus aureus organisms.

                                     ---- 

Product:          Milk and milk products under the Mathis Dairy label:
                  (a) Homogenized Whole Milk
                  (b) 2% Milk 
                  (c) Skim Milk 
                  (d) Chocolate Milk
                  (e) Buttermilk
                  (f) Half & Half 
                  (g) Whip Cream
                  (h) Egg Nog 
                  (i) Sun Up 2%.  Recall #F-175/183-1.
Code:             Buttermilk pull dates 1-27, 1-30 and all
                  other products pull dates:  1-8 and 1-11. 
Manufacturer:     Mathis Dairy, Decatur, Georgia
Recalled by:      Manufacturer by telephone January 2, 1991 followed by 
                  press release.  Firm-initiated recall complete. 
Distribution:     Georgia.
Quantity:         (a) 8732 gallons, (b) 8089 gallons, (c) 3798 gallons
                  (d) 1703 gallons, (e) undetermined, (f) 112 gallons,
                  (g) 185 gallons, (h) 1495 gallons; (i) undetermined, were 
                  distributed.

                                      -2- 





Reason:           Product is contaminated with alfatoxin. 

                                     ---- 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          Giant brand (a) Enriched White Sandwich Bread;
                  (b) Honey Wheat Bread, net wt 24 oz (1 lb 8 oz).
                  Recall #F-053/054-1.
Code:             THU on plastic tag. 
Manufacturer:     Giant Food, Inc., Silver Spring, Maryland.
Recalled by:      Giant Food, Inc., Landover, Maryland, by removing all 
                  merchandise from shelves, by contacting media and 
                  newspapers December 6, 1990 and by faxing list of 
                  affected products December 10, 1990.  Firm-initiated
                  recall complete.
Distribution:     Maryland, Virginia and Washington, D.C. 
Quantity:         (a) 19,310 loaves; (b) 51,660 loaves were distributed.
Reason:           Products were contaminated with sand. 

                                     ---- 

Product:          Lite Hand Cooked Gourmet Popcorn in 3/8 ounce metalized 
                  bags.  Recall #F-157-1. 
Code:             Lot numbers: JAN 11 JBC and JAN 11 JBDL.
Manufacturer:     Lite Snak, Inc., Randolph, Massachuetts.
Recalled by:      Boston Popcorn Company, Newton, Massachuetts by verbally
                  notifying delivery truck drivers to pick-up product from
                  distributors October 19, 1990.  Firm-initiated recall 
                  complete. 
Distribution:     Maine, New Hampshire, Connecticut, Massachuetts.
Quantity:         Firm estimates none remains on the market.
Reason:           Popcorn contained in bag labeled "microwave safe" 
                  disintegrates when microwaved.

                                     ---- 

Product:          Classic Milwaukee Brat Sauce in 18 ounce bottles: 
                  (a) Original; (b) Spicy; (c) Beer In flavors. 
                  Recall #F-196/198-1.
Code:             (a) H1080 A or B, H1270 A or B, H1560 A or B, H1570 A or B, 
                  H1670 A or B; 
                  (b) H1080 A or B, H1280 A or B, H1560 A or B; 
                  (c) H1090 A or B, H1280 A or B, H1570 A or B, H1660 A or B. 
Manufacturer:     Holsum Foods, Waukesha, Wisconsin.
Recalled by:      Manufacturer, by letter September 11, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Wisconsin, Illinois.

                                      -3- 





Quantity:         11,511 cases (12 bottles/case) were distributed.
Reason:           Product is contaminated with Lactobacillus bacteria.

                                     ---- 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 

                                     NONE 
                  Class II -

Product:          Red Blood Cells. Recall #B-103-1. 
Code:             Unit numbers:  2615622 and 2615629. 
Manufacturer:     Central Indiana Regional Blood Center, Indianapolis,
                  Indiana.
Recalled by:      Manufacturer, by telephone August 7, 1990.  Firm-initiated
                  recall complete.
Distribution:     Indiana.
Quantity:         2 units.
Reason:           Units drawn from donors living in malarial endemic areas
                  were distributed. 

                                     ---- 

Product:          Platelets, Pheresis.  Recall #B-106-1.
Code:             Unit #21P23374. 
Manufacturer:     American Red Cross Blood Services, Portland, Oregon.
Recalled by:      Manufacturer, by relabeling product August 9, 1990. 
                  Firm-initiated recall complete. 
Distribution:     Oregon. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.
                                     ---- 

Product:          Platelets, Pheresis.  Recall #B-108-1.
Code:             Unit #25P50063. 
Manufacturer:     American Red Cross Blood Services, Albany, New York.
Recalled by:      Manufacturer, by relabeling product June 21, 1990.
                  Firm-initiated recall complete. 
Distribution:     New York. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.

                                     ---- 

                                      -4- 





Product:          Platelets, Pheresis.  Recall #B-110-1.
Code:             Unit #02P05719. 
Manufacturer:     American Red Cross Blood Services, Wichita, Kansas. 
Recalled by:      Manufacturer, by relabeling product January 3, 1990.
                  Firm-initiated recall complete. 
Distribution:     Kansas. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.

                                    ----

                  Class III - 

Product:          Red Blood Cells.  Recall #B-104-1.
Code:             Unit #55E12071. 
Manufacturer:     American Red Cross, Little Rock, Arkansas.
Recalled by:      Manufacturer, by relabeling product April 9, 1990.
                  Firm-initiated recall complete. 
Distribution:     Arkansas. 
Quantity:         1 unit. 
Reason:           Blood product labeled with the incorrect expiration 
                  date was distributed. 

                                    ----

Product:          Whole Blood.  Recall #B-105-1.
Code:             Unit #PL1466. 
Manufacturer:     South Bend Medical Foundation, Inc., South Bend, Indiana. 
Recalled by:      Manufacturer, by telephone September 19, 1989.
                  Firm-initiated recall complete. 
Distribution:     Minnesota.
Quantity:         1 unit. 
Reason:           Unit of Whole Blood drawn in a collection bag containing
                  CPD anticoagulant/preservative solution (21 day expiry) 
                  but labeled as containing CPDA-1 anticoagulant/ 
                  preservative solution (35-day expiry) was distributed.

                                    ----

Product:          Red Blood Cells.  Recall #B-107-1.
Code:             Unit #31C06718. 
Manufacturer:     American Red Cross Blood Services, Buffalo, New York. 
Recalled by:      Manufacturer, by relabeling product July 18, 1990.
                  Firm-initiated recall complete. 
Distribution:     New York. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.

                                    ----


                                    -5- 





Product:          Red Blood Cells.  Recall #B-109-1.
Code:             Unit #42J59976. 
Manufacturer:     American Red Cross Blood Services, Cleveland, Ohio. 
Recalled by:      Manufacturer, by relabeling product July 20, 1990.
                  Firm-initiated recall complete. 
Distribution:     Ohio. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.

                                    ----

Product:          Recovered Plasma. Recall #B-111-1.
Code:             Unit Number: 894166.
Manufacturer:     Lake Charles Memorial Hospital, Lake Charles, Louisiana.
Recalled by:      Manufacturer,  by telephone September 11, 1990.  Firm-
                  initiated recall complete.
Distribution:     Texas, Florida. 
Quantity:         1 unit. 
Reason:           Unit which tested repeatably reactive for hepatitis B 
                  surface antigen (HBsAg) was distributed.

                                    ----

                       MEDICAL DEVICES AND RADIOLOGY

                  Class I - 

Product:          CPI Vista Pacemakers:  (a) Model 443 Implantable Pulse
                  Generator (single chamber); 
                  (b) Model 445 Implantable Pulse Generator (single 
                  chamber).  Recall #Z-194/195-1. 
Code:             Serial numbers: (a) 720875, 720876, 720878-720884,
                  720895, 720897-720903, 720943, 720944, 720959, 720961,
                  720906-720909, 720911-720914; 
                  (b) 721095, 721097-721110, 722404, 721876, 721877, 723477,
                  722375, 720925-720930, 721025-721032, 721041-721044,
                  722971, 722972, 721186, 721332-721338, 723955, 721339,
                  721450, 721452, 721453. 
Manufacturer:     Cardiac Pacemakers, Inc., St. Paul, Minnesota.
Recalled by:      Manufacturer, by letter December 21, 1990.
                  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         82 units were distributed.
Reason:           Pacemakers may exhibit early and rapid battery depletion
                  with a corresponding reduction in pulse amplitute.  In
                  addition, the potential exists for other devices in this
                  group to malfunction if the capacitor fails.

                                    ----

                                    -6- 





                  Class II -

Product:          Biflow nasal masks and cannulas:
                  (a) Comfort Cannula, product code 103000; 
                  (b) Biflow Nasal Mask with 7' Kink Resistant Tubing,
                  product code 103004;
                  (c) Comfort Cannula with 7' Kink Resistant Tubing,
                  product code 103005;
                  (d) Comfort Cannula with 14' Kink Resistant Tubing, 
                  product code 103015;
                  (e) Comfort Cannula with 25' Kink Resistant Tubing, 
                  product code 103025.  Recall #Z-215/219-1.
Code:             Lot numbers: 200016, 200122, 200298, 200299, 200423,
                  300401, 300696, 300697, 301220, 301221, 27898, 30188, 
                  32698, 300698;
                  (b) 200057, 200123, 200124, 200300, 200301, 200302, 
                  200303, 200424, 200425, 200426, 200427, 200607, 300402, 
                  300403, 300404, 300699, 301171, 301228, 301264, 27896,
                  27910, 29105, 32699, 34604; 
                  (c) 200058, 200059, 200304, 200305, 200428, 200429, 
                  200430, 200431, 200432, 200510, 300405, 300407, 300700, 
                  300701, 300702, 300703, 301205, 301222, 27899, 27911, 
                  28719, 29099, 30861, 34605, 35766;
                  (d) 200116, 200125, 301172, 301173, 27905, 35531, 35767;
                  (e) 200117, 200126, 200306, 200434, 200620, 301174, 
                  301202, 301203, 27900, 29958, 33393, 35768. 
Manufacturer:     Marquest Medical Products, Inc., Englewood, Colorado. 
Recalled by:      Manufacturer, by letter December 21, 1990 and 
                  January 4, 1991.  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         2,320 cases (50 units/case except item (e) which have 
                  25 units/case) were distributed.
Reason:           The product may have a restriction to gas flow impeding 
                  the delivery of supplemental oxygen.

                                    ----

Product:          Adult Nonconductive Disposable Anesthesia Circuit,
                  product code GM320L, an Rx single-patient use device for
                  gas and/or temperature monitoring.  Recall #Z-220-1.
Code:             Lot numbers: 300144, 300629, 300630, 300631.
Manufacturer:     Marquest Medical Products, Inc., Englewood, Colorado. 
Recalled by:      Manufacturer, by letter December 21, 1990.
                  Firm-initiated recall ongoing.
Distribution:     Texas, Louisiana. 
Quantity:         249 cases (20 units/case) were distributed. 
Reason:           The device was misassembled so that the inspiration 
                  and expiration limbs were reversed. 

                                    ----

                                    -7- 





Product:          Xenon rebreathing systems, single use devices that
                  serve as a conduit in administering Xenon-133 gas for 
                  steady state and washout lung perfusion diagnostic
                  studies:
                  (a) Xenon Rebreathing System with Mouthpiece and
                  Noseclip, Model #0133;
                  (b) Xenon Rebreathing System with Face Mask, Model #0134; 
                  Recall #Z-235/236-1.
Code:             Lot numbers: 90816A, 90921A, 90925B.
Manufacturer:     File Rite Products, Inc., Staten Island, New York 
                  (component).
Recalled by:      AMICI, Inc., Spring City, Pennsylvania, by letter 
                  on or about October 30, 1989.  Firm-initiated recall
                  complete. 
Distribution:     Nationwide. 
Quantity:         Firm estimates none remains on the market (270 systems
                  were returned). 
Reason:           A weakness in the neck of the bag can cause a tear, 
                  resulting in a Xenon leak.

                                    ----

Product:          LIFEPAK9 Defibrillator/Monitor, Part #803800- 
                  (all dash numbers).  Recall #Z-238-1. 
Code:             All lot numbers.
Manufacturer:     Physio-Control Corporation, Redmond, Washington.
Recalled by:      Manufacturer, by letter November 9, 1990.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         5,958 units were distributed. 
Reason:           Immediately after a defibrillation discharge, the ECG 
                  monitor may inappropriately display a flat trace
                  (resembling asystole) for a period of 10 seconds or 
                  longer when monitoring through the 3-lead patient cable.
                  This could result in misdiagnosis or temporary delay
                  in patient care.

                                    ----

Product:          LIFEPAK 10 Defibrillator/Monitor, part numbers affected 
                  are 804200-14, 804200-15, 804200-16, 804200-18, 804200-20,
                  804200-21, 804200-23.  Recall #Z-244-1. 
Code:             Lot numbers involved are noncontinuous starting with 6557 
                  and ending  with 7026:
                  Part No.      Serial No.
                  804200-14     006557, 006562, 
                  804200-15     007095 to 007104
                  804200-16     006957, 006968, 006969, 006972, 006973, 
                                006975 to 006979, 006981 to 006983, 
                                006986, 007080, 007083, 007085 to 007087, 
                  804200-18     006142, 006143, 006145, 006147, 006988, 
                                006992, 006994, 006999, 07000, 007004,
                                007006, 007011 to 007013, 007016, 

                                    -8- 





                  804200-20     006941 to 006947, 006949 to 06956, 07113 to 
                                007116, 007120 to 007124, 007128, 007130, 
                                007132, 
                  804200-21     007032 to 007034, 007036, 007037, 
                  804200-23     006302 to 006305, 006307 to 006310, 006312
                                to 006314, 006316 to 006323, 006325, 006327 
                                to 006330, 006333, 006335, 007018 to
                                007020, 007023, 007025, 007026. 
Manufacturer:     Physio Control Corporation, Redmond, Washington.
Recalled by:      Manufacturer, by letter November 12, 1990.  Firm-initiated
                  field correction ongoing. 
Distribution:     Nationwide and international. 
Quantity:         109 units are affected. 
Reason:           An incorrect component has been installed in the printed
                  circuit boards which have the potential to latch-up at
                  power on causing functions such as (sync, QRS, volume,
                  ECG size, lead select and default) to be inoperable.

                                    ----

Product:          Arrow-Howes Multi-Lumen Central Venous Catherization Kit, 
                  product #AK-15703-A, used principally to administer 
                  continuous or intermittent drugs, high volume or viscous
                  fluids, blood products or hyperalimentation nutrients to
                  seriously ill patients.  Recall #Z-252-1. 
Code:             Lot #AK-1-299-0.
Manufacturer:     Arrow International, Inc., Randleman, North Carolina. 
Recalled by:      Arrow International, Inc., Reading, Pennsylvania, by
                  telephone November 5 & 6, 1990 followed by letter 
                  November 9, 1990.  Firm-initiated recall complete.
Distribution:     Nationwide, Canada. 
Quantity:         3,155 kits were distributed; firm estimates none
                  remains on the market.
Reason:           The label states that the kits contain a 60 cm spring 
                  wire guide but may contain a 45 cm spring wire guide. 

                                    ----

Product:          FLEX 224 Resistor Trimmer, used for resistive trimming
                  of integrated circuits..  Recall #Z-253-1.
Code:             Serial numbers:  No. 1. 
Manufacturer:     Pacific Laser Corporation, Irvine, California.
Recalled by:      Manufacturer.  FDA approved the firm's corrective action
                  plan January 18, 1991.  Firm-initiated field correction 
                  ongoing.
Distribution:     California. 
Quantity:         1 unit. 
Reason:           Noncompliance with performance standard for laser products
                  in that the protective interlock system is inadequate.

                                    ----

                                    -9- 





Product:          Olympus Balloon Catheter Model B7-2C, designed for
                  endoscopic treatment of problems such as hemorrhage,
                  obstructions, collection of foreign bodies, and in x-ray
                  photography using contrast media.  Recall #Z-258-1. 
Code:             All lots. 
Manufacturer:     Olympus Optical Company, Ltd., Tokyo, Japan.
Recalled by:      Olympus Corporation, Lake Success, New York, by telephone 
                  on or about September 4, 1990.  Firm-initiated recall 
                  ongoing.
Distribution:     Utah, Florida, Georgia, New York. 
Quantity:         6 pieces were distributed.
Reason:           The device was marketed from May 1990 to September 1990 
                  without 510(k) approval.

                                    ----
                  Class III - 

Product:          IDS640 Integration Display & Storage Monitor, (in 
                  particular the Princeton Graphic Ultrasync 12 inch color
                  CRT Display Monitor Component).  Recall #Z-239-1. 
Code:             Serial numbers beginning with either "KF" or "KW".
Manufacturer:     Princeton Graphic Systems, Rosewell, Georgia (component). 
Recalled by:      Clinical Information Systems, Inc., Ann Arbor, Michigan,
                  by visit beginning March 1990.  Firm-initiated recall 
                  ongoing.
Distribution:     Michigan, New York, North Carolina, Illinois, 
                  California, Alabama.
Quantity:         110 systems were distributed. 
Reason:           The ultrasync display may go blank during use due to a
                  transistor failure. 

                                    ----

Product:          In-vitro Diagnostic Test Kit Activated AST Enzyme Set for 
                  use in the quantitative determination of AST activity in
                  serum or plasma.  AST determination is a liver function 
                  test: 
                  (a) Catalog No. 7070; (b) Catalog No. 8175 (bulk
                  packaging).  Recall #Z-245/246-1. 
Code:             Lot numbers:  (a) 1049 EXP 7/92, 1050 EXP 10/92 (Lot 1049 
                  contains lot 45 EXP 7/92 of Reagent Aspartate and Lot 47
                  EXP 10/92 of Reagent Enzyme;
                  (b) 1028 EXP 10/92 (Lot 1028 contains Lot 46A EXP 10/92 of
                  Reagent Aspartate and Lot 47A EXP 10/92 of Reagent
                  Enzyme).
Manufacturer:     AM Diagnostics (formerly American Monitor Corporation)
                  Indianapolis, Indiana.  Components of Lot 47 and 47A of 
                  Reagent Enzyme supplied by:  Lee Scientific, Inc.,
                  St. Louis, Missouri.
Recalled by:      AM Diagnostics, Indianapolis, Indiana, by telephone 
                  December 6 & 7, 1990 and by letter December 6, 1990.
                  Firm-initiated recall complete. 

                                    -10-





Distribution:     Nationwide and international. 
Quantity:         (a) 217 kits were distributed; firm estimates 139 kits
                  remain on the market; 
                  (b) 46 kits were distributed; firm estimates 23 kits
                  remain on the market. 
Reason:           Lots 47 and 47A of the enzyme reagent in the kits 
                  contain an above specification level of aspartate 
                  aminotransferase (AST).  This causes the Parallel, Excel
                  and Prospective Clinical Chemistry analyzers to show an 
                  alinear test response.

                                    ----

                            VETERINARY PRODUCTS 

                                    NONE