06/11/1991     

FDA Enforcement for the week of 6/12/91



                           FDA ENFORCEMENT REPORT 
                              FOR JUNE 12, 1991 

Recalls and Field Corrections:

                                    FOODS 

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                                      NONE

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 


Product:          Ceramic dinnerware sets that include dinner plates, salad 
                  plates, soup palates, serving bowls, and platters,
                  Pesaro pattern - (a) Dinner Plate; (b) Salad Plate; 
                  (c) Soup Plate; (d) 15" Bowl; (e) Platter;
                  Firenze pattern:  (f) Dinner Plate; (g) Salad Plate;
                  (h) Soup Plate; (i) 15" Bowl; (j) Two Handle Shallow Bowl;
                  (k) Platter.  Recall #F-387/397-1.

                                     -1-






Code:             Style numbers: (a) 88042; (b) 88040; (c) 88001; 
                  (d) 88032; (e) 88031; (f) 88041; (g) 88009; 
                  (h) 88043; (i) 88034; (j) 88005; (k) 88029. 
Manufacturer:     Maioliche il Bargello di Melani Anna Maria, Firenze, Italy. 
Recalled by:      Macy's California, Inc., San Francisco, California
                  (importer), by removing dinnerware sets from retail 
                  store shelves November 19, 1990 and by posting signs
                  in retail stores beginning February 1, 1991.  Firm- 
                  initiated recall ongoing. 
Distribution:     California. 
Quantity:         (a) 46 units; (b) 33 units; (c) 24 units; (d) 5 units;
                  (e) 6 units; (f) 48 units; (g) 36 units; (h) 25 units;
                  (i) 5 units; (j) 4 units; (k) 6 units were distributed. 
Reason:           High leachable levels of lead and cadmium in the product. 

                                     ---- 

Product:          Walgreen's A thru Z Jr. Children's Chewable Multi-
                  vitamin/Mineral Formula, in 60 tablet bottles.
                  Recall #F-399-1.
Code:             Item #505411, lot #9J0P EXP 8/91. 
Manufacturer:     Xcel Laboratories, Inc., Kalamazoo, Michigan. 
Recalled by:      Walgreen Company, Deerfield, Illinois, by electronic
                  messages April 19, 1990.  Firm-initiated recall ongoing.
Distribution:     Nationwide. 
Quantity:         6,650 bottles were distributed. 
Reason:           The product substituted cupric sulfate, which is an 
                  irritant to the mucous membrane, for cupric oxide.

                                     ---- 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          Snapple Orange Iced Tea, in 16 ounce glass bottles with 
                  metal screw cap and stick on label.  Recall #F-398-1. 
Code:             CT504222INV or T504222INZ printed on metal screw cap. 
Manufacturer:     Rumac Packaging Company, Trenton, New Jersey (bottler). 
Recalled by:      Snapple Natural Beverage Company, Ridgewood, New York,
                  by telephone September 4, 1990 followed by letter.
                  Firm-initiated recall complete. 
Distribution:     New York, New Jersey, Connecticut.
Quantity:         2,791 cases (24 bottles per case) were distributed. 
Reason:           Product has an off taste and had an oily substance. 

                                     ---- 

                                      -2- 





Product:          Liquid multi-vitamin supplement drops packed in 1 2/3 ounce 
                  containers: 
                  (a) Triple Vitamin Drops with Fluoride, 0.5 mg, an
                  Rx product; 
                  (b) Bio-Poly-Sol with Iron.  Recall #F-384/385-1. 
Code:             Lot numbers: (a) 004636 EXP 6/92; (b) 010632 EXP 12/92. 
Manufacturer:     Hi-Tech Pharmacal Company, Inc., Amityville, New York.
Recalled by:      Manufacturer.  Firm-initiated recall ongoing. 
Distribution:     Ohio. 
Quantity:         (a) 1,440 units; (b) 1,440 units were distributed;
                  firm estimates none remains on the market.
Reason:           A unit of Bio-Poly-Sol with Iron lot number 010632 was found
                  in a box labeled Triple Vitamin Drops with Floride 0.5 and a
                  unit of Triple Vitamin Drops with Floride 0.50 mg lot 004636
                  was found in a box labeled Bio-Poly-Sol with Iron.

                                     ---- 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 
                                     NONE 
                  Class II -

Product:          (a) Red Blood Cells; (b) Recovered Plasma.
                  Recall #B-231/232-1.
Code:             Unit numbers:  10H03604, 10J03977.
Manufacturer:     American Red Cross, Springfield, Missouri.
Recalled by:      Manufacturer, by letters of February 5 and 6, 1991
                  and by telephone December 19, 1990.  Firm-initiated recall
                  complete. 
Distribution:     Arkansas, Missouri, California. 
Quantity:         2 units of each component.
Reason:           Blood products collected from donors who reported use of
                  intravenous drugs or a history of hepatitis, were 
                  distributed.
                                     ---- 

Product:          Recovered Plasma.  Recall #B-246-1. 
Code:             Unit #21714.
Manufacturer:     Marietta Memorial Hospital, Marietta, Ohio. 
Recalled by:      Manufacturer, by letter March 18, 1990.  Firm-initiated 
                  recall complete.
Distribution:     Pennsylvania. 
Quantity:         1 unit. 

                                      -3- 





Reason:           Blood component, which tested non-reactive for
                  antibody to the human immunodeficiency virus type 
                  1 (anti-HIV-1) but collected from a donor who 
                  previously tested repeatably reactive for anti-HIV-1, 
                  was distributed.

                                     ---- 

Product:          Platelets, Pheresis.  Recall #B-249-1.
Code:             Unit #16P31275. 
Manufacturer:     American Red Cross, Columbus, Ohio. 
Recalled by:      Consignee notified the firm by telephone December 31, 
                  1989. Recall complete.
Distribution:     Ohio. 
Quantity:         1 unit. 
Reason:           Blood products labeled with incorrect expiration
                  dates, were distributed.

                                     ---- 

Product:          Source Plasma.  Recall #B-253-1.
Code:             Unit numbers: N16590-133 and N16652-133.
Manufacturer:     Hattiesburg Plasma Center, Inc., Hattiesburg, Mississippi.
Recalled by:      Manufacturer, during November 1990.  Firm-initiated 
                  recall complete.
Distribution:     California, Texas.
Quantity:         2 units.
Reason:           Source Plasma collected from a donor who participated 
                  in high-risk behavior was distributed.

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets; (c) Recovered 
                  Plasma.  Recall #B-265/267-1. 
Code:             Unit #27026-1742. 
Manufacturer:     United Blood Services, Division of Blood Systems
                  Inc., Texarkana, Texas. 
Recalled by:      Manufacturer, by telephone followed by letter 
                  December 5, 1990 and by letter of March 28, 1991. 
                  Firm-initiated recall ongoing.
Distribution:     Arkansas, North Carolina. 
Quantity:         1 unit of each component. 
Reason:           Blood components, collected from a donor who admitted 
                  to intravenous (IV) drug use, were distributed. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
                  Recall #B-268/270-1.
Code:             Unit #09R24827. 
Manufacturer:     American Red Cross, Omaha, Nebraska.

                                      -4- 





Recalled by:      Manufacturer, by letter July 13, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nebraska, California. 
Quantity:         1 unit of each component. 
Reason:           Blood products collected from a donor, who experienced
                  an occupational needlestick exposure two months prior 
                  to donation, were distributed.

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets; (c) Fresh Frozen
                  Plasma; (d) Recovered Plasma.  Recall #B-272/275-1. 
Code:             Unit numbers:  (a) 53T02522, 53L29642; (b) 53L29642;
                  (c) 53L29642; (d) 53T02522, 53G81925, 53H07803, 
                  53R21650, 53M49267. 
Manufacturer:     American Red Cross, Baltimore, Maryland.
Recalled by:      Manufacturer, by letters of February 15, 17 and 22,    1991.
                  Firm-initiated recall ongoing.
Distribution:     Maryland, Pennsylvania, California. 
Quantity:         (a) 2 units; (b) 1 unit; (c) 1 unit; (d) 5 units. 
Reason:           Blood products which tested non-reactive for
                  hepatitis B surface antigen (HBsAg), but were collected 
                  from donors who had previously tested repeatably
                  reactive for HBsAg, were distributed. 

                                     ---- 

Product:          Platelets.  Recall #B-276-1.
Code:             Unit #21S96001. 
Manufacturer:     American Red Cross Blood Services, Portland, Oregon.
Recalled by:      Consignee notified the firm on November 30, 1990. 
                  Recall complete.
Distribution:     Oregon. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.

                                     ---- 

Product:          Platelets.  Recall #B-277-1.
Code:             Unit #21S72085. 
Manufacturer:     American Red Cross Blood Services, Portland, Oregon.
Recalled by:      Manufacturer, by telephone April 11, 1990.  Firm- 
                  initiated recall complete.
Distribution:     Oregon. 
Quantity:         1 unit. 
Reason:           Blood product which tested positive for syphilis was
                  distributed for transfusion.

                                     ---- 

Product:          Platelets Pheresis.  Recall #B-281-1. 
Code:             Unit #41KP02993.

                                      -5- 





Manufacturer:     American Red Cross Blood Services, Huntsville, Alabama. 
Recalled by:      Consignee notified the firm, by telephone June 29, 1990.
                  Recall complete.
Distribution:     Alabama.
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration date
                  was distributed.

                                     ---- 

                  Class III - 

Product:          (a) Whole Blood; (b) Red Blood Cells; 
                  (c) Cryoprecipitated AHF; (d) Platelets; (e) Recovered
                  Plasma.  Recall #B-218/222-1. 
Code:             Unit numbers:  (a) 1033374; (b) 1028646, 2024453, 1268397;
                  (c) 1028646, 1268397; (d) 2024453; (e) 1028646, 2024453,
                  1268397.
Manufacturer:     LifeSource, Glenview, Illinois. 
Recalled by:      Manufacturer, by telephone March 18, 1991 and by
                  letter March 19, 1991.  Firm-initiated recall complete. 
Distribution:     Illinois, California. 
Quantity:         (a) 1 unit; (b) 3 units; (c) 2 units; (d) 1 unit; 
                  (e) 3 units.
Reason:           Blood components, which were collected from a donor 
                  who had previously tested repeatably reactive for the 
                  antibody to the Human Immunodeficiency Virus Type 1 
                  (anti-HIV-1) by EIA, were distributed.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-245-1.
Code:             Unit numbers: 45F58033, 45F58054, 45F58056, 45F58058, 
                  45F58059, 45F58062. 
Manufacturer:     American Red Cross Blood Services, Waterloo, Iowa.
Recalled by:      Manufacturer, by telephone January 4, 1991.  Firm-
                  initiated recall complete.
Distribution:     Iowa. 
Quantity:         6 units.
Reason:           Blood products, incorrectly tested for the antibody to
                  the hepatitis C virus (anti-HCV), were distributed for
                  transfusion.
                                     ---- 

Product:          Fresh Frozen Plasma.  Recall #B-247-1.
Code:             Unit numbers: 50G67323, 50G67341, 50G67344, 50G67346, 
                  50H73226. 
Manufacturer:     American Red Cross Blood Services, Toledo, Ohio.
Recalled by:      Consignee notified the manufacturer, by telephone 
                  July 3, 1990.  Recall complete. 
Distribution:     Ohio. 
Quantity:         5 units.

                                      -6- 





Reason:           Blood products, labeled with incorrect expiration dates,
                  were distributed for transfusion. 

                                     ---- 

Product:          Recovered Plasma.  Recall #B-248-1. 
Code:             Unit #01693.
Manufacturer:     Ephrata Community Hospital, Ephrata, Pennsylvania.
Recalled by:      Manufacturer, by letter April 3, 1991.  Firm-initiated
                  recall complete.
Distribution:     Pennsylvania. 
Quantity:         1 unit. 
Reason:           Blood component, which tested repeatably reactive for 
                  hepatitis B surface antigen (HBsAg) was distributed.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-250-1.
Code:             Unit #16H48551. 
Manufacturer:     American Red Cross, Columbus, Ohio. 
Recalled by:      Consignee notified the firm by telephone February 27, 1990. 
                  Recall complete.
Distribution:     Ohio. 
Quantity:         1 unit. 
Reason:           Blood product labeled with incorrect expiration 
                  dates, was distributed. 

                                     ---- 

Product:          Erasable-Programmable-Read-Only Memory (EPROM)
                  Computer Chips Numbers 15829-01, l5829-02, 15829-03,
                  15829-04, 15829-05 for use in the Haemonetics 
                  Plasmapheresis Collection System.  Recall #B-262-1. 
Code:             PCS Model 06001-CP - All serial numbers.
Manufacturer:     Haemonetics Corporation, Braintree, Massachusetts.
Recalled by:      Manufacturer, by telephone March 13, 1991 followed by 
                  letter March 14, 1991.  Firm-initiated field correction 
                  ongoing.
Distribution:     California, New York, Arizona, Illinois.
Quantity:         2,490 PCS's were distributed. 
Reason:           Defective computer chips caused no plasma separation
                  to occur during the first operating cycle.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-263-1.
Code:             Unit #401527. 
Manufacturer:     Belle Bonfils Memorial Blood Center, Denver, Colorado.
Recalled by:      Manufacturer, on March 6, 1991.  Firm-initiated 
                  recall complete.
Distribution:     Colorado. 
Quantity:         1 unit. 

                                      -7- 





Reason:           Blood component was distributed labeled with an 
                  incorrect expiration date.

                                     ---- 

Product:          Surgiscreen Reagent Red Blood Cells.  Recall #B-264-1.
Code:             Lot #3SS400, Cell #2 EXP 11/20/90.
Manufacturer:     Ortho Diagnostics Systems, Inc., Raritan, New Jersey. 
Recalled by:      Manufacturer, by letter November 2, 1990.  Firm-
                  initiated recall complete.
Distribution:     Nationwide. 
Quantity:         550 packages were distributed.
Reason:           Le(a+b+) Reagent Red blood Cells labeled as Le(a+b-)
                  were distributed. 

                                     ---- 

                         MEDICAL DEVICES AND RADIOLOGY

                  Class I - 

Product:          Software for Theraplan and TP-11 Cancer Treatment:
                  (a) Software Versions V048B-B and V05-PRx ("x" - 1, 2, 3
                  and 4) for Theratronics' Theraplan Treatment Planning 
                  Systems;
                  (b) Software Versions V08B-B and V09-RPx for Theratronics'
                  TP-11 Treatment Planning Systems.  Recall #Z-670/671-1. 
Code:             All units.
Manufacturer:     Theratronics International, Ltd., formerly Atomic Energy
                  of Canada, Ltd. (AECL Medical), Ontario, Canada.
Recalled by:      Theratronics International, Ltd., Carrollton, Texas,
                  by Urgent Software Bulletin November 23, 1990 and by
                  telephone November 26, 1990.  Firm-initiated field
                  correction ongoing. 
Distribution:     California, Maryland and international. 
Quantity:         97 disks were distributed.
Reason:           The software program may not accurately calculate the 
                  Cobalt 60 decay time and this may result in incorrect 
                  radiation treatment of the patient and therefore improper 
                  cancer treatment. 

                                     ---- 

                  Class II -

Product:          Theratron Phoenix Teletherapy (Cobalt 60) Unit, used
                  to treat cancer patients.  Recall #Z-722-1. 
Code:             All units.
Manufacturer:     Theratronics International, Ltd., formerly Atomic Energy
                  of Canada, Ltd. (AECL Medical), Ontario, Canada.
Recalled by:      Theratronics International, Ltd., Carrollton, Texas,
                  by User Bulletin CUB-90-3 dated July 3, 1990. 
                  Firm-initiated recall ongoing.

                                      -8- 





Distribution:     Nationwide and international. 
Quantity:         7 of 13 units distributed nationally have been
                  retrofitted.  39 units distributed internationally. 
Reason:           The single digital treatment timer on the Phoenix model 
                  is susceptible to ESD (Electrostatic discharge) which 
                  may cause timer failures resulting in a possible error
                  in radiation dose given to patients.

                                     ---- 

Product:          Fanuc Laser Models C-1000, C-1500, C-2000 and C-3000, 
                  used for industrial purposes.  Recall #Z-723/726-1. 
Code:             Serial numbers: Not determined. 
Manufacturer:     Fanuc, USA Corporation, Elk Grove Village, Illinois.
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan May 16, 1991.  Firm-initiated field 
                  correction ongoing. 
Distribution:     Nationwide. 
Quantity:         30 units were distributed.
Reason:           Noncompliance with performance standards for laser
                  products with respect to safety interlocks, instructions
                  for installation and system integration, and labeling for 
                  noninterlocked protective housings. 

                                     ---- 

Product:          Model MiniPro Laser Light Show Projector. 
                  Recall #Z-727-1.
Code:             Serial numbers:  None.
Manufacturer:     Laser Productions, Miami Beach, Florida.
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan May 16, 1991.  Firm-initiated field 
                  correction ongoing. 
Distribution:     Florida.
Quantity:         1 unit was distributed. 
Reason:           Noncompliance with performance standards for laser products 
                  in that it had inadequate output aperture masking, interlock
                  and aperture labels.

                                     ---- 

Product:          Companion 2801 Volume Ventilator, used for support for
                  ventilator dependant patients in the hospital home care 
                  center settings.  Recall #Z-728-1.
Code:             Serial numbers:  000102DE through 001173DE, 001246DE, 
                  000040PI, 000042PI, 000058PI. 
Manufacturer:     Puritan-Bennett, Portable Ventilator Division,
                  Boulder, Colorado.
Recalled by:      Manufacturer, by letter May 1, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, Canada, England, Puerto Rico, Argentina,
                  Italy.
Quantity:         740 units were distributed. 

                                      -9- 





Reason:           The internal battery can short cut, leading to potential
                  exposure of patient or caregiver to fumes; delay in alarm;
                  and system failure. 

                                     ---- 

Product:          Image Intensifier to OEC-Diasonics Models (a) DXR-5;
                  (b) DXR-10; (c) Model 902 Mobile C-Arm Imaging Systems. 
                  X-ray systems used in applications of fluoro imaging. 
                  Recall #Z-729/731-1.
Code:             Serial numbers: (a) 5C9001 through 5C9005, 5C9007 through 
                  5C9011, 5C9013 through 5C9034, 5CP6109, 5CP9/6048, 5CP9012, 
                  5CP9036 through 5CP9081, 5X6119, 5X9001 through 5X9017 and
                  5X9019 through 5X9024;
                  (b) 10C90002 through 10C9013, 10C9016 through 10C9019,
                  10CP9001, 10CP9020 through 10CP9064, 10X9001 and 10X9003
                  through 10X9014 and 10X9019;
                  (c) 2061, 2192/9, 2283, 2324CP, 9001 through 9047, 9049CP 
                  9050CP through 9053CP, A003 and A007. 
Manufacturer:     OEC-Diasonics, Inc., Salt Lake City, Utah.
Recalled by:      Manufacturer, by Field Service Bulletin (FSB) #90-017 dated 
                  10/3/90.  Firm-initiated field correction ongoing.
Distribution:     Nationwide, England, Japan, Germany, Poland, China, 
                  Canada, Equadore, Taiwan, Kenya.
Quantity:         Approximately 245 units were distributed. 
Reason:           The device was found to be hanging by partial unwound 
                  helicoils.
                                     ---- 

Product:          Dual Cuff Control Valve, for activating air pressure
                  to inflatable tourniquets used to create a bloodless
                  surgical field during surgery.  Recall #Z-732-1.
Code:             All lots of the product sold since January 1, 1987. 
Manufacturer:     A.B. Stille - Werner, Stockholm, Sweden.
Recalled by:      DePuy, Warsaw, Indiana, by letters of April 12 and 15, 1991 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide and Japan, Canada, Korea.
Quantity:         75 units were distributed.
Reason:           Premature wear to the internal valve components may cause 
                  detrimental leakage of pressure from the tourniquet cuff
                  during usage. 
                                     ---- 

Product:          Vertical Cassette Stand with automatic cassette centering 
                  and collimator sensing: 
                  (a) CS3 Model Nos. 51-761 and 51-779; 
                  (b) CS4 Model Nos. 55-424 and 55-432.  Recall #Z-733/734-1. 
Code:             Serial numbers (not in numerical order):
                  (a) 1001 - 9999 (Model #51-761), 345 - 9999 (Model 31-779); 
                  (b) 1002 - 1057 (Model 55-424), 1003 - 1023 (Model 55-432). 
Manufacturer:     Hans-Pausch Rontgengeratebau, GmbH, Erlangen, West Germany. 

                                     -10- 





Recalled by:      Siemens Medical Systems Inc., Iselin, New Jersey, by
                  memorandum during mid-1988.  Firm-initiated field correction
                  ongoing.
Distribution:     Nationwide, Puerto Rico, Canada.
Quantity:         (a) 317 units; (b) 24 units were distributed. 
Reason:           The cassette positioning bar may release inadvertently
                  and cause injury to patients or staff.

                                     ---- 

Product:          Abbott TestPack Plus Strep A Diagnostic Kit, a rapid
                  immunoassay for the qualitative detection of Group A
                  Streptococcal antigen from throat swab specimens or 
                  confirmation of presumptive positive Group A Streptococcal
                  colonies recovered from culture.  Recall #Z-736-1.
Code:             Lot numbers: 50065M100, 50383M200, 50454M200, 50517M100,
                  50539M300, 50539M301, 50539M302, 50550M300, 50551M300,
                  50551M301, 50552M300, 50553M300, 50554M300, 50768M100,
                  50776M100, 50887M200, 50931M200, 50931M201, 50931M202,
                  50932M200, 51050M100, 51051M100, 51052M100, 51053M100,
                  51061M300, 51679M100, 51679M101.
Manufacturer:     Abbott Laboratories, Diagnostics Division, Abbott Park, 
                  Illinois. 
Recalled by:      Manufacturer, by letter April 15, 1991 and by FAX 
                  April 22, 1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         3,747 20-test kits and 22,545  40-test kits were
                  distributed; firm estimates 20 percent remains on 
                  the market. 
Reason:           A potential exists for false positive results which 
                  exceeds the firm's current specificity claim. 

                                     ---- 

Product:          X-ray Controls and High Voltage Generators: 
                  (a) Model No. UD150B; (b) Model No. UD150K. 
                  Recall #Z-737/738-1.
Code:             All serial numbers. 
Manufacturer:     Shimadzu Corporation, Gardena, California.
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan May 6, 1991.  Firm-initiated field
                  correction ongoing. 
Distribution:     Nationwide. 
Quantity:         (a) 141 units; (b) 75 units were distributed. 
Reason:           Noncompliance with performance standards for x-ray products 
                  in that they do not have certification labels, model
                  number ID, date of manufacture, or product serial numbers.

                                     ---- 

Product:          (a) Sophy Camera DS7 System; (b) Sophy Camera DSX 
                  System.  Recall #Z-743/744-1. 

                                     -11- 





Code:             Serial numbers: (a) 313/1016 and above; 
                  (b) 301/1002 and above. 
Manufacturer:     Sopha Medical, France.
Recalled by:      Sopha Medical Systems, Inc., Columbia, Maryland, by 
                  letters of  April 12, 1991 and May 7, 1991. 
                  Firm-initiated field correction ongoing.
Distribution:     Nationwide and international. 
Quantity:         350 systems were distributed. 
Reason:           The detector/gantry assembly can mechanically travel beyond 
                  pre-set parameters or move in a fashion not expected by the 
                  operator. 

                                     ---- 

                  Class III - 

Product:          High Cholesterol Calibrator, a component of Kit 
                  Catalog #115641, used to determine serum cholesterol
                  levels.  Recall #Z-741-1. 
Code:             Lot numbers: K31JK, M49JK, D58LK. 
Manufacturer:     Ciba Corning Diagnostics Corporation, Oberlin, Ohio.
Recalled by:      Manufacturer, by letter dated February 15 and 18, 1991
                  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         1,303 kits were distributed.
Reason:           The high cholesterol calibrator (400mg/dl) has
                  a loss of activity that can result in elevated test values, 
                  causing inaccurate test results.

                                     ---- 

                              VETERINARY PRODUCTS 

                                     NONE