07/16/1991     

Enforcement Report for Week of July 17, 1991

                                                        July 17, 1991 
Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.

                                     NONE 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 

                                     NONE 

                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.

Product:          Best Health All Natural Gourmet Iced Tea, in 12 ounce 
                  bottles.  Recall #F-491-1.
Code:             D011 printed on the bottom of the bottle. 
Manufacturer:     Unknown.

                                      -1- 





Recalled by:      Miller Family of the Brooklyn Bottling Corporation, 
                  Brooklyn, New York, by telephone April 12, 1991,
                  followed by visit.  Firm-initiated recall complete. 
Distribution:     New York, New Jersey. 
Quantity:         1,400 cases.
Reason:           Product is contaminated with mold-like substance. 

                                     ---- 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 

                                     NONE 
                  Class II -

                       DEVICES AND RADIOLOGICAL PRODUCTS

                  Class I - 
                                     NONE 

                  Class II -

Product:          Bacitracin Differentiation Disk, an in-vitro diagnostic 
                  used to provide presumptive identification of group A 
                  Streptococci from other Lancefield groups of Hemolytic
                  Streptococci.  Recall #Z-674-1. 
Code:             Lot #110603 EXP 4/91. 
Manufacturer:     Organon Teknika, Durham, North Carolina.
Recalled by:      Manufacturer, by letter March 14, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Canada. 
Quantity:         9,816  50 disk cartridges were distributed; firm estimates
                  981 cartridges remain on the market.
Reason:           The product did not exhibit a zone of inhibition of 
                  growth with Streptococcus pyrogenes.

                                     ---- 

Product:          COBE 2991 Model 1 Cell Processor, equipped with LED 
                  Program Board Collect Valve, a centrifuge device that 
                  removes the preservative agent from a unit of red blood 
                  cells that has been frozen and subsequently thawed. 
                  Recall #Z-749-1.
Code:             Catalog numbers 912991-000, 912991-001, 912991-002, 
                  912991-005, 912991-010; Serial numbers 2000 through 2059. 
Manufacturer:     Cobe BCT, Inc., Blood Component Technology, Lakewood, 
                  Colorado. 

                                      -2- 





Recalled by:      Manufacturer, by telephone May 9, 1991, followed by 
                  letter June 6, 1991.  Firm-initiated field correction 
                  ongoing.
Distribution:     Nationwide, Canada, Australia, Belgium, Qatar, Portugal,
                  Switzerland, England, Japan, Spain. 
Quantity:         43 units were distributed.
Reason:           The centrifuge may stop spinning when the "CALL" or 
                  "ZERO SPIN TIME" functions are programmed.

                                     ---- 

Product:          Codman Holter Ventricular Catheter, Barium, Right Angle,
                  10 cm, indicated for use to gain access to the ventricles 
                  for diagnostic purposes and in the treatment of 
                  hydrocephalus.  Recall #Z-767-1.
Code:             Lot #GG-065, Model #82-1660.
Manufacturer:     Codman & Shurtleff, Inc., Randolph, Massachusetts.
Recalled by:      Manufacturer, by telephone March 5, 1991, followed by 
                  letter March 7, 1991 and by Telefax.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, Germany, England, France, Italy, Brazil,
                  Greece, Japan.
Quantity:         192 catheters were distributed. 
Reason:           10 cm catheters were incorrectly packaged and labeled 
                  as 7 cm catheters.
                                     ---- 

Product:          Gutta Percha Condenser Model 5003 (Touch N' Heat),
                  used in root canal procedures.  Recall #Z-768-1.
Code:             All units with serial numbers prior to 5003-5006. 
Manufacturer:     Analytic Technology Corporation, Redmond, Washington. 
Recalled by:      Manufacturer, by letter August 29, 1990.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         980 units were distributed. 
Reason:           The heating unit may not turn off.  The unit may behave 
                  as if the touch switch was permanently engaged.  The
                  problem appears to have been caused by a bad phenloic 
                  insulator bushing located inside the probe. 

                                     ---- 

Product:          Kendall Curity Spinal Anesthesia Tray, Reorder
                  Code 4706.  Recall #Z-770-1.
Code:             Lot #DT8349KG EXP 9/30/91.
Manufacturer:     Kenmex Especialidades Medicar, Tijuana, Mexico. 
recalled by:      Kendall Healthcare Products Company, Mansfield, 
                  Massachusetts, by letter May 7, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         120 cases (10 trays per case) were distributed. 

                                      -3- 





Reason:           The trays may contain expired drug components, i.e.,
                  Marcaine, Adrenalin and Ephedrine.

                                     ---- 

Product:          Intravenous transfer tubes used mainly by oral
                  surgeons and dentists to administer anesthesia during 
                  oral surgery: 
                  (a) Anesthesia Extension Tube,
                  Model No. A-12, Order No. 003-0012
                  (b) Anesthesia Extension Tube,
                  Model No. A-12M, Order No. 003-0121 
                  (c) Anesthesia Extension Tube,
                  Model A-20, Order No. 003-0020
                  (d) Anesthesia Extension Tube,
                  Model No. A-20M, Order No. 003-0201 
                  (e) Anesthesia Extension Tube,
                  Model No. A-30, Order No. 003-0030
                  (f) Anesthesia Extension Tube,
                  Model No. A-30M, Order No. 003-0301 
                  (g) Administration Set, 
                  Model AE-60, Order No. 003-0060 
                  (h) Administration Set, 
                  Model AE-60V, Order No. 003-0061
                  (i) Administration Set, 
                  Model A-78, Order No. 003-0078
                  (j) Venoset, Order No. 03-1127
                  (k) Irrigation Set, Model IR-60, Order No. 003-1621 
                  (l) Irrigation Set, Model IR-60V, Order No. 003-1620. 
                  Recall #Z-771/782-1.
Code:             Lot numbers 2827 through 2836.
Manufacturer:     Ace Surgical Supply, Brockton, Massachusetts. 
Recalled by:      Manufacturer, by telephone June 7, 1991 followed by 
                  letter.  Firm-initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         (a) 69 cases; (b) 79 cases; (c) 583 cases; (d) 281 cases; 
                  (e) 236 cases; (f) 217 cases; (g) 23 cases; (h) 53 cases; 
                  (i) 37 cases; (j) 185 cases; (k) 9 cases; (l) 4 cases were
                  distributed.
                  (The first 6 products are shipped in cases of 250 units 
                  per case; the last six products are shipped in cases of 
                  50, the firm stated that some of these last 6 products
                  are sold individually.) 
Reason:           Questionable sterility because of package design. 

                                     ---- 

Product:          Rasp Adaptor 10 mm from the HG Multilock Hip Prosthesis.
                  Recall #Z-797-1.
Code:             Catalog #6631-010-02, Lot #7578100. 
Manufacturer:     Tuco, Inc., Instrument Division, Canton, Massachusetts. 

                                      -4- 





Recalled by:      Zimmer, Inc., Warsaw, Indiana, by telephone April 23, 1991
                  followed by letter April 24, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         36 pieces were distributed. 
Reason:           The rasp adaptor will not lock into the corresponding 
                  rasp handle.

                                     ---- 

Product:          DelMed Brand Y-Blood Solution Set with Large Filter,
                  a medical device restricted to sale by or on order of 
                  a physician, used for mixing, infusing or reinfusing
                  patients.  Recall #Z-798-1. 
Code:             Catalog #03-122-00, lot numbers: 8J-020-04, 8K-006-04;
                  9B-006-04X. 
Manufacturer:     DelMed, Inc. S.A., Soyapango San Salvador, El Salvador. 
Recalled by:      CharterMed, Inc., Lakewood, New Jersey, by letter 
                  May 14, 1991.  Firm-initiated recall ongoing. 
Distribution:     New York, California, Virginia, Massachusetts, Illinois,
                  Washington, D.C.
Quantity:         265 cases (48 units/case) were distributed. 
Reason:           Lots 8J-020-04 and 8K-006-04 may contain defective roller 
                  clamps which may break during use, resulting in interrupted 
                  or unrestricted flow of solution through the tubing; and
                  lot 9B-006-04X may have loosened needle adaptor protectors, 
                  resulting in an inadequate sterility barrier. 

                                     ---- 

Product:          MIII Mammography System Bucky Adapters. 
                  Recall #Z-799-1.
Code:             All serial numbers. 
Manufacturer:     Lorad Medical Systems, Inc., Danbury, Connecticut.
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan June 14, 1991.  Firm-initiated field correction 
                  ongoing.
Distribution:     Nationwide. 
Quantity:         16 units were distributed.
Reason:           Noncompliance with performance standards for x-ray
                  products in that the bucky adaptor changes the size of
                  the image receptor without changing the aperture. 
                  This may result in the x-ray field not being aligned
                  properly with the image receptor. 

                                     ---- 

Product:          Schneider Teflon Fixed Core Guidewires, .038 inches.
                  Recall #Z-801-1.
Code:             Lot #18802. 
Manufacturer:     Schneider (USA), Inc., Pittsburgh, Pennsylvania.

                                      -5- 





Recalled by:      Schneider (USA) Inc., Plymouth, Minnesota, by 
                  telephone April 18, 1991.  Firm-initiated recall ongoing. 
Distribution:     Florida, New Jersey, Pennsylvania, Spain. 
Quantity:         400 were distributed. 
Reason:           The seals on the sterile packages may not be secure.

                                     ---- 

Product:          Chemspore Chemical and Biological Indicators, packaged
                  with 15 ampules per carton, Part #P-76427-015,
                  Catalog #NA016 or with 50 ampules per carton, 
                  Part #P-764271-017, Catalog #NA018.  Recall #Z-802-1. 
Code:             Lot numbers: CS031E EXP 3/93, CS041B EXP 4/93,
                  CS041C EXP 4/93.
Manufacturer:     AMSCO Scientific, Apex, North Carolina. 
Recalled by:      Manufacturer, by letter May 7, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, Canada. 
Quantity:         269 cartons were distributed. 
Reason:           The manufacturer changed the media and the lot of media 
                  used is killing the spores, thus changing resistance
                  of the indicator. 

                                     ---- 

Product:          Strato Micropump (a) Model 2400 Strato Micropump
                  Ambulatory Medication Infuser, Catalog #8810332;
                  (b) Model 2500 Strato Micropump Programmer, 
                  Catalog #8810331 and 8810335 (international version). 
                  Recall #Z-804/805-1.
Code:             All serial numbers. 
Manufacturer:     Strato Medical Corporation, Beverly, Massachusetts. 
Recalled by:      Manufacturer, by telephone April 17, 1991 followed by 
                  letters of April 17, 1991 and May 2, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide, Germany, Canada, Spain. 
Quantity:         (a) 55 units; (b) 59 units were distributed.
Reason:           Two of the three operational modes do not operate 
                  properly due to software errors.  Overinfusion of 
                  medication is possible under certain circumstances. 
                  The effect of improper therapy would be dependant on the
                  type and dosage of drug infused through the infuser.

                                     ---- 

Product:          Lemon Glycerine Pre-Moistened Swabs:
                  (a) Triad Lemon Glycerine Insta-Swab, Reorder #10-4002; 
                  (b) ABCO Lemon Gylcerine Swabs, Reorder #5424500. 
                  Recall #Z-813/814-1.
Code:             Lot numbers: (a) 1D76, 1D84; (b) 1D92.
Manufacturer:     H & P Industries, Inc., Franklin, Wisconsin.

                                      -6- 





Recalled by:      Manufacturer, by telephone May 29, 1991 and by letter 
                  June 5, 1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         649 cases (10 cartons/cases, 25 pouches/carton) were
                  distributed.
Reason:           The above specified lots may have been contaminated 
                  with Pseudomonas microorganisms.

                                     ---- 

CORRECTION:       Fetal Bovine Serum (FBS), Recall #Z-763/764-1 which 
                  appeared in the June 19, 1991, Enforcement Report should
                  read: 

                  Reason:  The Fetal Bovine Serum (FBS) has been implicated 
                  in being imported illegally into the United States labeled
                  as human serum from Brazil. 

                                     ---- 

                  Class III - 

Product:          Johnson & Johnson K-Y Lubricating Jelly, in 4 ounce tubes.
                  Recall #Z-803-1.
Code:             All lots of 4 ounce tubes in blister packs with the 
                  reorder number 8919 printed on the tube.
Manufacturer:     Johnson & Johnson Consumer Products, Inc., Skillman,
                  New Jersey. 
Recalled by:      Merchandise, Inc., Miamitown, Ohio, by letter 
                  December 24, 1990.  Firm-initiated recall complete. 
Distribution:     Ohio, Kentucky, Indiana, Illinois, Michigan, New York,
                  Florida, Pennsylvania.
Quantity:         136 tubes were distributed; firm estimates none 
                  remains on the market.
Reason:           Product is not labeled appropriately for retail use.

                                     ---- 

                              VETERINARY PRODUCTS 

                                     NONE 

Seizure Actions Filed:

Product:          Surgical Instruments (91-577-526).
Charge:           Adulterated - Quality falls below that which they 
                  are represented to possess in that they are represented 
                  as stainless steel surgical instruments, but do not contain 
                  sufficient amount of chromium.
                  Misbranded - Labeling is false and misleading in that 
                  they are labeled as stainless steel suggesting that the 
                  devices contain a certain amount of chromium, when in 
                  fact they do not contain such amount of chromium. 

                                      -7- 





Firm:             Tamzon Instruments, Inc., Northbrook, Illinois. 
Filed:            July 3, 1991 - U.S. District Court for the Northern 
                  District of Illinois; Civil #91C4128; FDC #66208. 

                                     ---- 

Product:          Surgical Instruments (91-577-413/414).
Charge:           Adulterated - Quality falls below that which they 
                  are represented to possess in that they are represented 
                  as stainless steel surgical instruments, but do not contain 
                  sufficient amount of chromium.
                  Misbranded - Labeling is false and misleading in that 
                  they are labeled as stainless steel suggesting that the 
                  devices contain a certain amount of chromium, when in 
                  fact they do not contain such amount of chromium. 
Firm:             Magnum Instruments, Inc., Addison, Illinois.
Filed:            July 3, 1991 - U.S. District Court for the Northern 
                  District of Illinois; Civil #91C4129; FDC #66211. 

                                     ---- 

Product:          Surgical Instruments (91-577-610-416).
Charge:           Adulterated - Quality falls below that which they 
                  are represented to possess in that they are represented 
                  as stainless steel surgical instruments, but do not contain 
                  sufficient amount of chromium.
                  Misbranded - Labeling is false and misleading in that 
                  they are labeled as stainless steel suggesting that the 
                  devices contain a certain amount of chromium, when in 
                  fact they do not contain such amount of chromium. 
Firm:             Surgical Instruments, Inc., Chicago, Illinois.
Filed:            July 3, 1991 - U.S. District Court for the Northern 
                  District of Illinois; Civil #91C4130; FDC #66212. 

                                     ---- 

Product:          Love Me Slender with Enhancer 6000 Guar Gum (91-378-283). 
Charge:           Product is an unapproved new drug and its labeling lacks
                  adequate directions for use.
Firm:             Ad Fill Service, doing business as Love Me Slender, 
                  Las Vegas, Nevada.
Filed:            May 9, 1991 - U.S. District Court for the District of 
                  Nevada; Civil #CV-S-91-354 LDG-LRL, FDC #66085. 

                                     ---- 

                                      -8- 





Product:          QLF-1 Frequency Generator (91-573-941). 
Charge:           Misbranded - Labeling is false and misleading and fails 
                  to bear adequate directions for use.
Firm:             Pat Hamm, M.D., Huntsville, Alabama.
Filed:            May 17, 1991 - U.S. District Court for the Northern 
                  District of Alabama, Northeastern Division; 
                  Civil #CV-91-HM-1213-NE, FDC #66156.

                                     ---- 

Product:          Frozen Shrimp (90-608-325). 
Charge:           Product is decomposed and contains rodent hair. 
Firm:             Cumberland Foods, Nashville, Tennessee. 
Filed:            October 19, 1990 - U.S. District Court for the
                  Middle District of Tennessee, Nashville Division; 
                  Civil #3 90 0919, FDC #65946.