FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 Enforce
08/06/1991    

Enforcement Report for August 7, 1991 


                                                       August 7, 1991 

Recalls and Field Corrections:

                                     FOODS

                  Class I - A situation in which there is a reasonable
                  probability that the use of, or exposure to, a violative
                  Product will cause serious adverse health consequences or 
                  death.


Product:          Butter Reddies, portion control butter service for use
                  by the food service industry, packed in unlabeled trays 
                  containing either 90 or 72 patties per tray, and sold in
                  15 pound cartons.  Paper label on top of butter patties 
                  says Bell or Main Street.  Recall #F-500-1. 
Code:             150, 151, 157, 158 stamped on shipping case only. 
Manufacturer:     Cabarrus Creamery Company, Concord, North Carolina
                  (bulk butter); Kay's Ice Cream Company, Knoxville, Tennessee
                  (Cream).
Recalled by:      Dairy Sales, Coatesville, Pennsylvania, by telephone
                  June 28, 1991, followed by letter July 9, 1991. 
                  Firm-initiated recall ongoing.
Distribution:     Pennsylvania. 

                                      -1- 






Quantity:         1,211 cartons.
Reason:           Product manufactured without heat treatment using cream 
                  contaminated with Listeria monocytogenes. 

                                     ---- 

Product:          Bulk Butter in 68 pound cubes.  Recall #F-507-1.
Code:             100-143-01 through 100-143-05.
Manufacturer:     Cabarrus Creamery Company, Inc., Concord, North Carolina. 
Recalled by:      Manufacturer, by telephone June 28, 1991.  Firm-initiated 
                  recall complete.
Distribution:     Pennsylvania. 
Quantity:         271 bulk cubes were distributed.
Reason:           Product was manufactured without heat treatment using cream 
                  contaminated with Listeria monocytogenes. 

                                     ---- 

                  Class II - A situation in which the use of, or exposure to
                  a violative product may cause temporary or medically
                  reversible adverse health consequences or where the 
                  probability of serious adverse health consequences is 
                  remote. 


Product:          C J's Treats, Juice Filled Stix, Waxed Jumbo Stix, packed 
                  in 7/8 ounce bags.  Recall #F-503-1.
Code:             All packages not identified as containing FD&C Yellow #5. 
Manufacturer:     Glenn Confection Division, Buffalo, New York. 
Recalled by:      M & M Distributing Company, La Plume, Pennsylvania, by
                  visit beginning March 3, 1991.  Firm-initiated recall 
                  complete. 
Distribution:     Pennsylvania. 
Quantity:         Approximately 975 packages were distributed.
Reason:           Product contains undeclared FD&C Yellow #5. 

                                     ---- 

Product:          Fresh Yellow Fin Tuna (five fish headed and gutted, 
                  each fish weighing 107 pounds, 151 pounds, 120 pounds,
                  101 pounds and 110 pounds).  Recall #F-506-1. 
Code:             None. 
Manufacturer:     Imported from Ecuador.
Recalled by:      Poseidon Enterprises, Inc., Charlotte, North Carolina,
                  by telephone May 24 and 28, 1991.  Firm-initiated recall
                  ongoing.
Distribution:     North Carolina, South Carolina. 
Quantity:         Approximately 400 pounds were distributed.
Reason:           Product is decomposed.

                                     ---- 

                                      -2- 





                  Class III - A situation in which the use of, or exposure
                  to a violative product is not likely to cause adverse 
                  health consequences.


Product:          Barcelona Foods brand dried fruit and nut mixes 
                  in plastic bags:
                  (a) 3 ounce Rainey Lake Fruit Bowl; 
                  (b) 3 ounce Rainey Lake Fruit and Fiber;
                  (c) 3 ounce I Love Fruit 'N Nuts; 
                  (d) 2-3/4 ounce and 3 ounce Fruit 'N Nuts;
                  (e) 3 ounce Rainey Lake I Love Fruit 'N Nuts; 
                  (f) 3 ounce Tropical Trail Mix; 
                  (g) 3-1/2 ounce Just Fruit; 
                  (h) 3-1/4 ounce Rainey Lake Honey Sweet;
                  (i) 3 ounce Rainey Lake Tropical Rainbow. 
                  Recall #F-429/437-1.
Code:             Not coded.
Manufacturer:     Barcelona Nut Company, Baltimore, Maryland. 
Recalled by:      Manufacturer, by placing stickers on products to be 
                  completed by March 22, 1991.  Firm-initiated field
                  correction complete.
Distribution:     Virginia, Washington, D.C., Connecticut, New Jersey,
                  Massachusetts, Maryland.
Quantity:         Firm estimates all units were corrected.
Reason:           Product labeling failed to declare the presence 
                  of an artificial color. 

                                     ---- 

Product:          Slush Puppy Ice Poppies, in 1 ounce and 2 ounce plastic 
                  sheaths, flavored ice items.  Recall #F-501-1.
Code:             The shipping cases are identified with date, 4-9-91.
Manufacturer:     American Food Products, Inc., Cincinnati, Ohio. 
Recalled by:      Manufacturer, by telephone May 20, 1991, followed by visit. 
                  Firm-initiated recall ongoing.
Distribution:     Ohio, Pennsylvania, Mississippi.
Quantity:         Approximately 5,220 of the 24 count cartons and 100 
                  cases of the 2 ounce product were distributed; firm 
                  estimates none remains on the market. 
Reason:           Product is contaminated with yeast. 

                                     ---- 

Product:          Flamingo brand Assorted Tropical Fruit Nectars in 10
                  ounce cans.  Recall #F-502-1. 
Code:             All lots
Manufacturer:     Multi Fruit C.A., Yaracuy, Venezuela. 
Recalled by:      Blanditos Food, Miami, Florida, by telephone November 6,
                  1990.  Firm-initiated recall ongoing. 
Distribution:     Florida.
Quantity:         1,980 cases were distributed. 

                                      -3- 





Reason:           Product contained in defective cans.

                                     ---- 

Product:          Sunnyfresh Citrus Punch Drink.  Recall #F-504-1.
Code:             SELL BY MAY 22. 
Manufacturer:     Juiceland, Inc., Detroit, Michigan. 
Recalled by:      Manufacturer, by visit.  Firm-initiated recall complete.
Distribution:     Michigan. 
Quantity:         407 cases (8 bottles per case) were distributed.
Reason:           Product labeled as containing 68 ounces but packaged
                  in 64 ounce bottles.

                                     ---- 

                                   COSMETICS

                                     NONE 

                           HUMAN DRUGS AND BIOLOGICS

                  Class I - 

                                     NONE 

                  Class II -

Product:          Clenbuterol Tablets, 20 mcg, in 100 tablet bottles, an
                  illegal veterinary product which is being marketed as a 
                  human anabolic steroid alternative.  Recall #D-513-1. 
Code:             All lots. 
Manufacturer:     Pacifico Enterprises, Dayton, Ohio (distributor). 
Recalled by:      Distributor, by letter July 8, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         78 bottles were distributed.
Reason:           Product is an unapproved new drug.

                                     ---- 

Product:          Recovered Plasma.  Recall #B-347-1. 
Code:             Unit numbers LZ01636, LZ01673.
Manufacturer:     National Health Guard, Inc., Fort Lauderdale, Florida.
Recalled by:      Manufacturer, by telephone March 21, 1991, and by 
                  letter March 22, 1991.  Firm-initiated recall ongoing.
Distribution:     Florida.
Quantity:         2 units.
Reason:           Recovered Plasma, which tested repeatably reactive
                  for the antibody to the hepatitis C virus (anti-HCV)
                  and which was distributed for further manufacture into
                  non-injectable products, was not labeled as reactive
                  for anti-HCV. 
                                     ---- 

                                      -4- 





Product:          Recovered Plasma.  Recall #B-349-1. 
Code:             Unit numbers:  1582375, 1588059, 1600092. 
Manufacturer:     Palm Beach Blood Bank, Inc., West Palm Beach, Florida.
Recalled by:      Manufacturer, by letter April 4, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Florida.
Quantity:         3 units.
Reason:           Blood products, which tested repeatably reactive for the
                  antibody to the hepatitis C virus (anti-HCV), but 
                  labeled as non-reactive for anti-HCV, were distributed. 

                                     ---- 

Product:          (a) Red Blood Cells; (b) Platelets.  Recall #B-350/351-1. 
Code:             Unit numbers: (a) 374770, 424478; (b) 424478. 
Manufacturer:     Belle Bonfils Memorial Blood Center, Denver, Colorado.
Recalled by:      Manufacturer, by letter February 6, 1991.  Firm-
                  initiated recall complete.
Distribution:     Colorado. 
Quantity:         (a) 2 units; (b) 1 unit.
Reason:           Blood components, which were: 1) not tested for the 
                  antibody to the human immunodeficiency virus, type 1
                  (anti-HIV-1); or 2) improperly tested for the antibody to 
                  the human T lymphotropic virus, type 1 (anti-HTLV-1); or
                  3) tested initially reactive for anti-HTLV-I, were
                  distributed.
                                     ---- 

Product:          (a) Red Blood Cells; (b) Recovered Plasma.
                  Recall #B-353/354-1.
Code:             Unit #1029105.
Manufacturer:     Logan Regional Hospital, Logan, Utah. 
Recalled by:      Intermountain Health Care Blood Services, Salt Lake 
                  City, Utah, by letters of May 20 and 23, 1991.
                  Firm-initiated recall complete. 
Distribution:     Utah, Florida, New Jersey, Massachusetts. 
Quantity:         1 unit of each component. 
Reason:           Blood products, which tested initially reactive for the 
                  antibody to the human immunodeficiency virus type 1 
                  (anti-HIV-1), were distributed. 

                                     ---- 

Product:          Platelets, Pheresis.  Recall #B-355-1.
Code:             Unit #272456. 
Manufacturer:     Community Blood Center, Dayton, Ohio. 
Recalled by:      Consignee discovered labeling error and notified the
                  recalling firm on January 15, 1991.  Recall complete. 
Distribution:     Ohio. 
Quantity:         1 unit. 
Reason:           Blood product labeled with an incorrect expiration
                  date was distributed. 

                                     ---- 
                                      -5- 





Product:          Whole Blood.  Recall #B-356-1.
Code:             Unit #W11120. 
Manufacturer:     Hendrick Medical Center, Abilene, Texas.
Recalled by:      Manufacturer, by telephone November 30, 1990.  Firm-
                  initiated recall ongoing. 
Distribution:     Texas.
Quantity:         1 unit. 
Reason:           Blood product which tested repeatably reactive for
                  the antibody to the human immunodeficiency virus type 1 
                  (anti-HIV-1), was distributed for transfusion.

                                     ---- 

Product:          Source Plasma.  Recall #B-357-1.
Code:             Unit numbers: N-39698, N-39812, N-39881, N-40014, N-40098,
                  N-39655.
Manufacturer:     San Juan Plasma, Inc., Farmington, New Mexico.
Recalled by:      Manufacturer, by telephone April 30, 1991.  Firm- 
                  initiated recall ongoing. 
Distribution:     California. 
Quantity:         6 units.
Reason:           Source Plasma, which tested non-reactive for the
                  hepatitis B surface antigen (HBsAg) and the antibody to 
                  the human immunodeficiency virus type 1 (anti-HIV-1) but
                  was collected from a donor who previously tested reactive 
                  for either HBsAg or anti-HIV-1, was released. 

                                     ---- 

                  Class III - 

Product:          ProSom Estazolam Tablets, 2 mg, in 100 tablet bottles,
                  an Rx oral Schedule IV controlled drug for short-term 
                  management of insomnia.  Recall #D-511-1. 
Code:             Lot numbers: 48-895-AF-21, 48-895-AF-22.
Manufacturer:     Abbott Laboratories, Abbott Park, Illinois. 
Recalled by:      Manufacturer, by telephone July 19, 1991, followed by 
                  letter July 23, 1991.  Firm-initiated recall ongoing. 
Distribution:     Illinois, Wisconsin, Michigan, Ohio, Indiana, Kentucky. 
Quantity:         264 bottles were distributed; firm estimates 75 percent 
                  of the product remains on the market. 
Reason:           Some of the tablets bear the identification code "UC" 
                  instead of the correct code "UD". 

                                     ---- 

Product:          Heparin Sodium Injection USP, 1000 USP units/ml, in 
                  30 ml vials, an Rx injectable for intravenous and 
                  subcutaneous use for anticoagulant therapy. 
                  Recall #D-512-1.
Code:             Catalog #01930, lot #910525 EXP 11/92.
Manufacturer:     Smith & Nephew SoloPak, Elk Grove Village, Illinois.
Recalled by:      Manufacturer, by telephone July 18, 1991, followed by 
                  letter July 22, 1991.  Firm-initiated recall ongoing. 

                                      -6- 





Distribution:     New Jersey, Nevada, Washington state, Arizona, Florida, 
                  Oklahoma, California, Tennessee, Oregon, Arkansas.
Quantity:         360 cases (25 vials per case) were distributed; firm
                  estimates 316 cases remain on the market. 
Reason:           Superpotency. 
                                     ---- 

Product:          Vivotif Berna Typhoid Vaccine, Live Oral Ty21a. 
                  Recall #B-287-1.
Code:             Lot #115382A EXP 10/91. 
Manufacturer:     Swiss Serum and Vaccine Institute, Berne, Switzerland.
Recalled by:      Berna Products Corporation, Coral Gables, Florida, by 
                  letter, telephone and FAX February 11, 1991.  Firm- 
                  initiated recall ongoing. 
Distribution:     Nationwide. 
Quantity:         18,900 units were distributed.
Reason:           Product was misbranded in that it was distributed 
                  containing an unapproved package insert.

                                     ---- 

Product:          Red Blood Cells.  Recall #B-294-1.
Code:             Unit numbers MWBB 61992, MWBB 62007, MWBB 62021,
                  MWBB 62984. 
Manufacturer:     Metropolitan Washington Blood Banks, Inc., Silver Spring, 
                  Maryland. 
Recalled by:      Consignee returned three units for relabeling December 26,
                  1990 and 1 unit had been transfused.  Recall complete.
Distribution:     Maryland and Washington, D.C. 
Quantity:         4 units.
Reason:           Blood products collected from a donor who tested
                  initially reactive for the antibody to the human
                  immunodeficiency virus type-1 (anti-HIV-1); or
                  labeled with wrong expiration dates, were distributed.

                                     ---- 

Product:          Recovered Plasma.  Recall #B-352-1. 
Code:             Unit numbers: 374770, 424478, 481642. 
Manufacturer:     Belle Bonfils Memorial Blood Center, Denver, Colorado.
Recalled by:      Manufacturer, by letter February 6, 1991 and by 
                  telephone March 8, 1991 followed by letter April 5, 1991. 
                  Firm-initiated recall complete. 
Distribution:     Pennsylvania, California, Missouri. 
Quantity:         3 units.
Reason:           Blood components, which were: 1) not tested for the 
                  antibody to the human immunodeficiency virus, type 1
                  (anti-HIV-1); or 2) improperly tested for the antibody to 
                  the human T lymphotropic virus, type 1 (anti-HTLV-1); 
                  or 3) tested initially reactive for anti-HTLV-1, were 
                  distributed.
                                     ---- 

                                      -7- 





                       DEVICES AND RADIOLOGICAL PRODUCTS

                  Class I - 
                                     NONE 

                  Class II -

Product:          Theratron Cobalt 60 Teletherapy Unit: 
                  (a) Theratron 60 Cobalt 60 Teletherapy Unit (T60);
                  (b) Theratron 80 Cobalt 60 Teletherapy Unit (T80),
                  includes units refurbished to original specifications 
                  which are identified by Theratronics as T80R's);
                  (c) Theratron 780 Cobalt 60 Teletherapy Unit (T780).
                  Recall #Z-694/696-1.
Code:             All T60 and T80 units and Serial Numbers 1 through 8 for
                  T780 units. 
Manufacturer:     Theratronics International, Ltd., formerly Atomic Energy of 
                  Canada, Ltd (AECL Medical), Ontario, Canada.
Recalled by:      Theratronics International, Ltd., Carrollton, Texas,
                  by CUB - 90-2 dated July 7, 1990.  Firm-initiated recall
                  ongoing.
Distribution:     Nationwide and international. 
Quantity:         455 units were distributed. 
Reason:           The spring pins may collapse and shear which would
                  allow the gantry to rotate freely.

                                     ---- 

Product:          (a) Software Versions V01, V02 and V03 for Theratronics'
                  Theraplan treatment planning system with central
                  processor PDP 11/34 and 200 Series; 
                  (b) Software Versions V01, V02, V03, V04, V05, V06 and
                  V07 for Theratronics' TP-11 treatment planning
                  systems.  Recall #Z-720/721-1.
Code:             All units.
Manufacturer:     Theratronics International, Ltd., formerly Atomic 
                  Energy of Canada, Ltd. (AECL Medical), Kanata, Ontario
                  Canada (components).
Recalled by:      Theratronics International, Ltd., Carrollton, Texas,
                  by a document titled "Software Hazard Report" dated 
                  March 30, 1987.  Firm-initiated field correction ongoing. 
Distribution:     Nationwide. 
Quantity:         252 software versions were distributed. 
Reason:           The computer prompt may request units of radiation activity 
                  of mgm units instead of the required computation of mCi 
                  dosage units.  This may result in an error in the 
                  radiation treatment given to the patient. 

                                     ---- 

Product:          SLT Fiber Optic Adaptor, used for various surgical
                  procedures.  Recall #Z-735-1. 
Code:             Serial numbers: Undetermined. 

                                      -8- 





Manufacturer:     Ximed Medical Systems, Santa Clara, California. 
Recalled by:      Manufacturer.  FDA approved the firm's corrective action plan 
                  May 24, 1991.  Firm-initiated field correction ongoing. 
Distribution:     Unknown.
Quantity:         3 units were distributed and were returned to manufacturer. 
Reason:           Noncompliance with the performance standards for laser
                  products in that the device may defeat the interlocks 
                  on the SLT laser system.

                                     ---- 

Product:          Search N Stim Electronic Acupuncture (device), model
                  SW-103, used to help reduce pain and improve vitality.
                  Recall #Z-757-1.
Code:             All lots. 
Manufacturer:     Marco Polo Industries & Merchandising Company, LTD.,
                  Kowloon, Hong Kong. 
Recalled by:      Damark International, Inc., Minneapolis, Minnesota, by
                  letter dated May 23, 1991.  Firm-initiated recall complete. 
Distribution:     Nationwide. 
Quantity:         48 units were distributed.
Reason:           Unapproved medical device sold to general public by mail
                  order.
                                     ---- 

Product:          Lasercut HP Laser Cutting Systems.  Recall #Z-783-1.
Code:             Serial numbers 101 through 117. 
Manufacturer:     Lasercut Inc., North Branford, Connecticut. 
Recalled by:      Manufacturer.  FDA approved the firm's corrective 
                  action plan June 19, 1991.  Firm-initiated field
                  correction ongoing. 
Distribution:     Pennsylvania, Massachusetts, Ohio, Connecticut, 
                  North Carolina, New Jersey, New York, Tennessee,
                  Iowa. 
Quantity:         17 units were distributed.
Reason:           Noncompliance with the performance standard for laser 
                  products in that they failed to have warning logotypes, 
                  certification, aperture, and noninterlocked protective
                  labels.  Also, the "Caution-use of controls..." statement 
                  was not included in the sales brochure. 

                                     ---- 

Product:          Coilcath Hemodialysis Catheters:
                  (a) CVC 300 Coilcath Dual Lumen Hemodialysis Catheter,
                  40 cm length; 
                  (b) CVC 305 Coilcath Dual Lumen Hemodialysis Catheter with
                  introducer kit, 40 cm length; 
                  (c) CVC 310 Coilcath Dual Lumen Hemodialysis Catheter,
                  36 cm length; 
                  (d) CVC 315 Coilcath Dual Lumen Hemodialysis Catheter 
                  with introducer kit, 36 cm length;

                                      -9- 





                  (e) CVC 301 Coilcath Catheter Extension with injection
                  caps.  Recall #Z-808/812-1. 
Code:             All lot numbers.
Manufacturer:     Strato Medical Corporation, Beverly, Massachusetts. 
Recalled by:      Manufacturer, by telephone May 2, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         275 units were distributed. 
Reason:           Product is mislabeled.  Also the firm's investigational 
                  device exemption (IDE) submission has not been approved 
                  by the Food and Drug Administration.

                                     ---- 

Product:          Autotransfusion kits and accessories: 
                  (a) Product AT2000 Kit, in cardboard outer box colored
                  blue and white labeled in part "Impraflo autotransfusion
                  kit AT 2000 for whole blood salvage"; 
                  (b) Product AT2100, paratubing set which is plastic tubing
                  for blood suctioning; 
                  (c) Product AT2120 ACD-A: Bottle of Anticoagulant;
                  (d) Product AT2200 Hanger: Attachment for IV pole 
                  (e) Product AT3510 Kit:  Components in two trays, 
                  canister and tubing trays;
                  (f) Product AT3515: Same as 3510 except two drain tubes 
                  are 3/16";
                  (g) Product AT3519 Kit:  Same as 3510 except two wound
                  drain tubes are 1/4"; 
                  (h) Product AT3600 Kit: Contains 600 ml drainage/ 
                  reinfusion canister with discard adapter in a tray
                  sterilized for single use.  Recall #Z-815/822-1.
Code:             All lots. 
Manufacturer:     Impra, Inc., Tempe, Arizona.
                  Cytosol Laboratories, Inc., Braintree, Massachusetts
                  (Product AT 2120).
Recalled by:      Impra Inc., Tempe, Arizona, by letter January 14, 1991, 
                  and by telephone January 8, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide, Germany, France, United Kingdom,
                  Switzerland, Belgium. 
Quantity:         (a) 2,410 units; (b) 723 units; (c) 845 units;
                  (d) 280 units; (e) 59 units; (f) 168 units; (g) 21 units; 
                  (h) 88 units were distributed.
Reason:           Documentation deficiencies were noted in the
                  manufacturing process.

                                     ---- 

Product:          Intra-Aortic Balloon Catheters Mansfield Sidewinder SL
                  (Single Lumen) and Sidewinder DL (Double Lumen), used in
                  the circulatory assistance of patients with compromised 
                  blood pumping capability: 

                                     -10- 





                  (a) Catalog #3296 (with Introducer DL Kit Catalog #3442), 
                  DL, 9.5 F, 40 cc; 
                  (b) Catalog #3296-NC (with Introducer DL Kit Catalog
                  #3443-NC), DL, 9.5 F, 40 cc guidewire not heparin coated; 
                  (c) Catalog #3299 (with Insertion Kit Catalog #3452), 
                  DL, 9 F, 30 cc; 
                  (d) Catalog #3140 SL, 10 F, 40 cc.  Recall #Z-823/826-1.
Code:             Lot numbers:  (a) 3781, 36667, 36673, 35491, 35490, 36492,
                  32184, 900668, 31634, 30605, 29912, 25040, 21726, 21640,
                  20831, 20139, 20140, 18959, 18699, 18435, D050A, D030N, 
                  C140H;
                  (b) 33473, 35110, 900526, 900529; 
                  (c) 35534, 33908; 
                  (d) 36309, 33236, 31633, 31388. 
Manufacturer:     Boston Scientific Corporation, Mansfield, Massachusetts.
Recalled by:      Boston Scientific Corporation, Watertown, Massachusetts, by 
                  telephone June 7, 1991, followed by letters of June 13 and
                  14, 1991.  Firm-initiated recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         Approximately 2,220 units were distributed.  FDA estimates
                  50 percent remains on the market. 
Reason:           The balloon may not unfurl completely when inflation is 
                  attempted.
                                     ---- 

Product:          Multiflex Modules known as Slow TimeShare (STS) 
                  Units, used for switching products for feeding laser output 
                  from one industrial workstation to another and for welding
                  and drilling applications.  Recall #Z-830-1.
Code:             Undetermined. 
Manufacturer:     Luminics Ltd., Norwichshire, England. 
Recalled by:      Luminics Corporation, Livonia, Michigan.  FDA approved the
                  firms corrective action plan June 28, 1991.  Firm-initiated 
                  field correction ongoing. 
Distribution:     Nationwide. 
Quantity:         22 units were distributed.
Reason:           Noncompliance with performance standard for laser 
                  products in that assembly wear was discovered resulting 
                  in incomplete shutter closure.

                                     ---- 

Product:          Airborne Neonatal Transport Incubator, Model 20-H 
                  equipped with 4" Rhombus Casters, Part #367A32C100WZLD37
                  (without brakes) and Part #367H32C100WZLD37 (with 
                  brakes).  Recall #Z-869-1.
Code:             Serial Numbers  0197XF, 0199XF, 0200XF, 0201XF, 0204XF, 
                  0205XF, 0206XF, 0215XF, 0202XG, 0208XG, 0209XG, 0210XG, 
                  0211XG, 0203XH, 0207XH, 0212XH, 0213XH, 0214XH, 0216XH, 
                  0217XJ, 0219XK, 0220XK, 0221XK, 0223XK, 0224XK, 0225XL, 
                  0226XL, 0227XL, 0230XL, 0231XL, 0232XL, 0233XL, 0222XM, 
                  0228XM, 0229XM, 0234XM, 0235YA. 
Manufacturer:     Rhombus Casters, Inc., Marlton, New Jersey. 

                                     -11- 





Recalled by:      International Biomedical, Inc., Austin, Texas, by 
                  letter January 22, 1990.  Firm-initiated recall complete. 
Distribution:     Nationwide. 
Quantity:         257 casters were distributed.  148 of these casters 
                  were used for initial manufacture of 37 incubators, 
                  the remainder were sold as stock replacements.
Reason:           The casters can unexpectedly disengage from the air/
                  oxygen cart and separate. 

                                     ---- 

Product:          Catheter Port Valves, used for maintaining a fluid-tight
                  seal around a dilitation catheter during angioplasty: 
                  (a) Rotating Hemostatic Valve, Catalog #23242;
                  (b) Standard Hemostatic Valve, Catalog #23240;
                  (c) Duostat Rotating Hemostatic Valve, Catalog #23244;
                  (d) Guide Wire Accessory Kit (contains a Rotating Hemostatic
                  Valve), Catalog #22295.  Recall #Z-872-1. 
Code:             Lot numbers: (a) 9110751 through 9122251 and 0010251 through
                  0030851; (b) 9110351 through 9122251 and 0010951
                  through 0011051; (c) 9110751 through 9122251 and
                  0010251 through 0030851; (d) 9110751 through 9122251 and
                  0010251 through 0030851.
Manufacturer:     Advanced Cardiovascular Systems, Inc., Temecula,
                  California. 
Recalled by:      Manufacturer, by letter and by FAX April 3, 1991. 
                  Firm-initiated recall ongoing.
Distribution:     Nationwide and international. 
Quantity:         69,664 units were distributed; firm estimates none
                  remains on the market.
Reason:           The affected valves may not unseal when opened.  This 
                  failure to unseal can result in the failure of the
                  valve to properly flush air prior to use or can 
                  restrict the introduction/movement of the catheter
                  system through the valve. 

                                     ---- 

Product:          Buchwald Wigness Central Venous Access Catheters: 
                  (a) Buchwald Wigness (50 cm) Single Lumen Bidirectional 
                  Valved Catheter and 8 French Introducer Kit, Model WBC 100; 
                  (b) Buchwald Wigness (65 cm) Single Lumen Bidirectional 
                  Valved Catheter and 8 French Introducer Kit, Model WBC 105. 
                  Recall #Z-876/877-1.
Code:             Lot numbers:  (a) 016405, 019416, 010711; 
                  (b) 015502, 012403, 012005. 
Manufacturer:     Strato Medical Corporation, Beverly, Massachusetts. 
Recalled by:      Manufacturer, by telephone September 27, 1990.
                  Firm-initiated recall ongoing.
Distribution:     Unknown.
Quantity:         266 units were distributed. 

                                     -12- 





Reason:           Pressure exceeding the recommended maximum of 25 psi may
                  result in the small tubing valve of the catheter popping
                  out of the inlet valve.  In addition, the firm had shipped
                  some units to Italy in May 1990, prior to receiving 510(k)
                  approval to market (September 21, 1990), and without the
                  appropriate export certificate. 

                                     ---- 
                  Class III - 

Product:          Kodak Ektachem Clinical Chemistry Slides (TBIL), an 
                  in-vitro diagnostic slide for use with Kodak Ektachem 
                  Analyzers to quantitatively measure total bilirubin 
                  (TBIL).  Recall #Z-871-1. 
Code:             Catalog #154 2570, Lot #4204-0138-3469. 
Manufacturer:     Eastman Kodak Company, Clinical Products Division,
                  Rochester, New York.
Recalled by:      Manufacturer, by letter June 3, 1991.  Firm-initiated 
                  recall ongoing. 
Distribution:     Nationwide. 
Quantity:         Firm estimates 12 sodium cartridges packed in TBIL
                  cartons remain on the market. 
Reason:           Some sodium cartridges were inadvertently packaged in TBIL
                  boxes, i.e., boxes labeled as Clinical Chemistry Slide
                  (TBIL) may actually contain sodium slides.

                                     ---- 

Product:          Kodak Ektachem (a) 500 Series; (b) 700 Series Analyzers 
                  with (Product) PCS II accessory unit.  Recall #Z-878/879-1. 
Code:             All lot numbers for Catalog numbers 8351058, 8407520, 
                  1943083, 8195653, 8005290, 1687763, 1434851.
Manufacturer:     Eastman Kodak Company, Clinical Products Division,
                  Rochester, New York.
Recalled by:      Manufacturer, by letter April 3, 1991.  Firm-initiated
                  recall ongoing. 
Distribution:     Nationwide and international. 
Quantity:         1,064 were distributed in the United States and 50-60 
                  were distributed internationally. 
Reason:           Some of the disposable tips may cause false low CO(2) 
                  or falsely high AST results.

                                     ---- 

                              VETERINARY PRODUCTS 

                  Class I - 
                                     NONE 
                  Class II -

Product:          E-Tex Range (700) Medicated, Anaplasmosis Free Choice 
                  Medicated Block, 33-1/3 pounds, for ruminants only. 
                  Recall #V-075-1.

                                     -13- 





Code:             Product manufactured during the months of March and 
                  April 1991. 
Manufacturer:     East Texas Feeds, Inc., Mineola, Texas. 
Recalled by:      Manufacturer, by telephone May 6, 1991.  Firm-initiated 
                  recall complete.
Distribution:     Texas.
Quantity:         572 blocks were distributed; firm estimates none remains
                  on the market.
Reason:           Free Choice Medicated Blocks is a new animal drug and 
                  was being manufactured and distributed without the required 
                  approved New Animal Drug Application. 

                                     ---- 

Product:          Kent Flav-R-Ized 20% Pig Nuggets A150 Medicated with
                  Apramycin, in 50 pound bags.  Recall #V-076-1.
Code:             Product #2348, lot #R 36. 
Manufacturer:     Kent Feeds, Inc., Rockford, Illinois. 
Recalled by:      Manufacturer, by telephone June 18, 1991 and
                  July 12, 1991.  Firm-initiated recall ongoing.
Distribution:     Illinois, Wisconsin.
Quantity:         405 bags were distributed; firm estimates 5 bags remain 
                  on the market.
Reason:           Feed was subpotent in Apramycin.

                                     ---- 

Medical Device Safety Alerts: 

Product:          Kodak Surecell Chlamydia Test Kits, an in-vitro 
                  diagnostic test kit for use in a doctor's office
                  or clinical lab, for the diagnosis of Chlamydia:
                  (a) Kodak Surecell Chlamydia Test Kits, Catalog #193 7804;
                  (b) Kodak Surecell Chlamydia test Kits, Catalog #825 0003;
                  (c) Kodak surecell Chlamydia Test Kits, Catalog #196 8056.
                  Safety Alert #M-033/035-1.
Code:             All lots. 
Manufacturer:     Eastman Kodak Company, Biological Diagnostics Clinical
                  Products Division, Rochester, New York. 
Alerted by:       Manufacturer, by letters of  April 8 and 11, 1991.
Distribution:     Nationwide and international. 
Quantity:         28,970 kits were distributed; firm estimates 55 percent 
                  of the kits remain on the market. 
Reason:           There is a potential for false positive results using 
                  some Type I swabs due to an interaction with the swab 
                  adhesive and Kodak Surecell Chlamydia Test Kit. 

                                     ---- 
Seizure Actions Filed:

Product:          Neomycin, an animal drug (91-597-322).
Charge:           Lack of current good manufacturing practices and
                  failure to drug list. 

                                     -14- 





Firm:             RX Veterinary Products, Kansas City, Missouri.
Filed:            July 2, 1991 - U.S. District Court for the Western
                  District of Missouri, Western Division; 
                  Civil #91-0594-CV-W-6, FDC #66193.

                                     ---- 

Product:          Steviasweet Diet Tea, Assorted Flavors (91-331-516).
Charge:           Adulterated - Product contains a food additive which is 
                  unsafe and because there is no regulation prescribing the 
                  conditions under which the additive may be safely used. 
Firm:             Steviasweet Company, Inc., Arlington, Texas.
Filed:            June 4, 1991 - U.S. District Court for the Northern 
                  District of Texas, Fort Worth Division; Civil #4-91-406-E,
                  FDC #66081.