FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
 
  
  
  
  October 1, 1997                                      97-40
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
  _______________
  PRODUCT      Veragel 1:1 Purified Aloe Vera Leaf Water
                 Soluble Extract, in 55 blue gallon
                 polyethylene drums.  Recall #F-822-7.
  CODE         Lot #6-1781, Product Code 21501-000, No
                 expiration date.
  MANUFACTURER Madis Botanicals, South Hackensack, New
                 Jersey.
  RECALLED BY  Manufacturer, by telephone on November 13,
                 1996.  Firm-initiated recall complete.
  DISTRIBUTION Nationwide.
  QUANTITY     57 drums were distributed; firm estimates none
                 remains on the market.
  REASON       The product is contaminated with Enterobacter
                 gergoviae, an opportunistic pathogen.
  
  _______________
  PRODUCT      Blue-Marlin Frozen Loins, individually wrapped
                 in plastic and packed into a cardboard case.
               Recall #F-824-7.
  CODE         Lot #20140.
  MANUFACTURER Costa Pesca, Quito, Ecuador
                 (shipper/processor).
  RECALLED BY  Fresh Fish Inc., Miami, Florida, by letter
                 faxed on June 10, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION Florida and Maryland.
  QUANTITY     1,303 pounds were distributed.
  REASON       Product was associated with scombroid
                 poisoning and contained high histamine levels.
  
 
  
  ____________
  PRODUCT      "The" Daily Multiple Tablets, shipped to the
                 repacker in a bulk fiber drum.
               Recall #F-825-7.
  CODE         Bulk drum: lot #960135;  Unit bottle: lot
                 #C96008.
  MANUFACTURER Bactolac Pharmaceuticals, Inc., Westbury, New
                 York.
  RECALLED BY  Manufacturer, by letter mailed on August 15,
                 1996, followed by telephone.  Firm-initiated
                 recall complete.
  DISTRIBUTION Illinois.
  QUANTITY     Approximately 60,000 tablets were distributed;
                 firm estimates none remains on the market.
  REASON       Product is subpotent in folic acid.
  
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
  _______________
  PRODUCT      Tatum's and Country Hearth brands, Old Time
                 Biscuits, made with buttermilk, net weight 8
                 ounces.      Recall #F-826-7.
  CODE         Pull date of 07/07/97 on bag.
  MANUFACTURER Tatum Bakers, Birmingham, Alabama.
  RECALLED BY  Manufacturer, by visit.  Firm-initiated recall
                 complete.
  DISTRIBUTION Florida.
  QUANTITY     3,260 packages were distributed.  Firm
                 estimates none remains on the market.
  REASON       Product is unfit for food due to the presence
                 of aluminum fragments.
  
  ______________
  PRODUCT      Cashews under the Thrifty brand and D&W brand:
                 a) Whole Cashews, 2-1/4 ounces; b) Cashews
                 (broken kernels), 8 ounces; c) Salted Cashews,
                 2-3/4 ounces.  Recall #F-827/829-7.
  CODE         All product with expiration date of 12/97 or
                 earlier.
  MANUFACTURER Thrift Products Company, Kentwood, Michigan.
  RECALLED BY  Manufacturer, by telephone on June 9, 1997,
                 followed by visit on June 10 and 11, 1997, and
                 by press releases on June 9 and 10, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Michigan.
  QUANTITY     Approximately 7,500 bags were distributed.
  REASON       The products' ingredient statement does not
                 list the source of the oil used for roasting.
  
                               -2-_______________
  PRODUCT      Minute Maid Orange Juice packaged in 9-packs
                 of 8.45 ounce (250 ml.) aseptic drink box
                 containers. Unit count per case is 27.
               Recall #F-832-7.
  CODE         Lot number: Package: MAR0698GF
               Case: XX:XX MAR0698 G20581D.
  MANUFACTURER The Minute Maid Company, Geneva, Ohio.
  RECALLED BY  The Minute Maid Company, Houston, Texas, by
                 telephone on September 12, 1997, followed by
                 visit.  Firm-initiated recall ongoing.
  DISTRIBUTION Pennsylvania, New York, Ohio, Alabama,
                 Mississippi.
  QUANTITY     1,383 cases were distributed.
  REASON       Product is contaminated with yeast.
  
  _______________
  UPDATE       Recall #F-812-7, Red Sugar Cake Decoration,
                 which appeared in the September 24, 1997,
                 Enforcement Report is being amended to NOT
                 INCLUDE the Betty Crocker 1 ounce product. 
                 The Kroger Brand Red Sugar Cake Decoration is
                 correctly included under Recall #F-812-7.
  
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS I ============
  _______________
  PRODUCT      Prograf Tacrolimus Capsules, 1 mg, in bottles
                 of 100, Rx oral drug indicated for the
                 prophylaxis of organ rejection in patients
                 receiving allogenic liver transplants.  
               Recall #D-276-7.
  CODE         Lot 1C3062B EXP 4/30/99.
  MANUFACTURER Fujisawa Ireland Ltd., Republic of Ireland.
  RECALLED BY  Fujisawa USA, Inc., Deerfield, Illinois, by
                 letters dated September 3 and 5, 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     6,890 bottles were distributed; firm estimated
                 that 25-30% of product remained on market at
                 time of recall initiation.
  REASON       Mislabeling -- At least one bottle labeled to
                 contain 1 mg capsules actually contained 5 mg
                 capsules.
  
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
  _______________
  PRODUCT      Gentamicin Sulfate Injection, USP, 40 mg/mL,
                 in 20 ml multiple dose vials, under the
                 following labels:  Steris, Rugby, Goldline
                 (Zenith on label).       Recall #D-277-7.
  CODE         Lot #95K550.
                               -3-MANUFACTURER   Steris Laboratories, Inc., Phoenix, Arizona.
  RECALLED BY  Manufacturer, by letter on March 31, 1997.
                 Firm-initiated. 
  DISTRIBUTION Nationwide.
  QUANTITY     19,000 vials were distributed.
  REASON       White precipitate in product (preservatives).
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
  _______________
  PRODUCT      Platelets.  Recall #B-1389-7.
  CODE         Unit #16337-9672.
  MANUFACTURER Blood Systems, Inc., Meridian, Mississippi.
  RECALLED BY  Manufacturer, by letter dated April 8, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Mississippi.
  QUANTITY     1 unit was distributed.
  REASON       Blood product corresponded to a unit of Red
                 Blood Cells that was reported as clotted.
  
  _______________
  PRODUCT      Allergenic Extract and Allergenic Treatment
                 Sets.  Recall #B-1412-7.
  CODE         Lot # 648304, 653265, 663176, 666156, 668804,
                 668925, 668926, 670458, 671334, 678222,
                 678223, 678224, 678225, 678226, 678227,
                 678228, 678229, 678230, 678231, 678232,
                 678233, 678234, 678235, 678236, 678237,
                 678238, 678239, 678240, 678241, 692263,
                 693755.
  MANUFACTURER ALK Laboratories, Berkeley, California.
  RECALLED BY  Manufacturer, by letter on August 16 and 27,
                 1997, followed by telephone.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     31 lots.
  REASON       Lack of assurance that products met sterility
                 testing requirements prior to release.
  
  _______________
  PRODUCT      Platelets, Pheresis.  Recall #B-1465.
  CODE         Unit numbers:  16333-9425 and 16333-9430,
  MANUFACTURER Blood Systems, Inc., Meridian, Mississippi.
  RECALLED BY  Manufacturer, by letter dated October 23,
                 1996.  Firm-initiated recall complete.
  DISTRIBUTION Mississippi.
  QUANTITY     2 units were distributed.
  REASON       Blood products may have been labeled with an
                 extended expiration date.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1468-7.
  CODE         Units 49FJ53975, 49FJ55367, 49FM26720.
                               -4-MANUFACTURER   American National Red Cross, Tulsa, Oklahoma.
  RECALLED BY  Manufacturer, by letters dated July 1, 1997,
                 and August 4, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION Oklahoma and Texas.
  QUANTITY     3 units were distributed.
  REASON       Blood product was collected from a donor who
                 reported travel to an area designated as 
                 endemic for malaria. 
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Fresh Frozen Plasma. 
                 Recall #B-1472/1473-7.
  CODE         Unit 29271-0984.
  MANUFACTURER United Blood Services, Chicago, Illinois.
  RECALLED BY  Manufacturer, by telephone on May 29, 1997,
                 and by letter dated July 8, 1997.  Firm-initiated recall complete.
  DISTRIBUTION Illinois.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 with a history of cancer. 
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets.  
               Recall #B-1474/1475-7.
  CODE         Unit numbers:  a) 10005367, 13000623,
                 31006566, 31006572; b) 31006566, 31006572.
  MANUFACTURER W.E. & Lela I. Stewart Blood Center, Inc,
                 Tyler, Texas.
  RECALLED BY  Manufacturer, by telephone on March 19, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     a) 4 units; b) 2 units were distributed.
  REASON       Blood products tested repeatedly reactive for
                 the Hepatitis B Surface Antigen (HBsAg), were
                 distributed. 
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Fresh Frozen Plasma.
                 Recall #B-1492/1493-7.
  CODE         Unit numbers:  a) 10161-2218 and 10158-5733;
                 b) 10156-1459, 10161-2218,            10158-5733.
  MANUFACTURER United Blood Services,    Scottsdale, Arizona.
  RECALLED BY  Blood Systems, Inc.,               Scottsdale, Arizona, by
                 letters dated July 5, 1996, and December 10,
                 1996.  Firm-initiated recall complete.
  DISTRIBUTION Arizona and California.
  QUANTITY     a) 2 units; b) 3 units were distributed.
  
                               -5-REASON         Blood products tested negative for the
                 antibody to the human T-lymphotropic virus
                 type I (anti-HTLV-I), but were collected from
                 a donor who previously tested repeatedly
                 reactive for anti-HTLV-1.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1494-7.
  CODE         Unit #10525-2279.
  MANUFACTURER Blood Systems, Inc., Scottsdale, Arizona.
  RECALLED BY  Manufacturer, by telephone on February 28,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Arizona.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was prepared from a unit of
                 Whole Blood with an extended collection time.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1496-7.
  CODE         Unit #M96019058.
  MANUFACTURER United Blood Services, Ventura, California.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 telephone on June 20, 1996.  Firm-initiated
                 recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected in a manner that
                 could compromise the sterility of the product.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1497-7.
  CODE         Unit #4661234.
  MANUFACTURER Naval Hospital, Groton, Connecticut.
  RECALLED BY  Manufacturer, by telephone on August 28, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Haiti.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested repeatedly reactive for
                 the antibody to the hepatitis B core antigen
                 (anti-HBc).
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III  =====
  _______________
  PRODUCT      a) Red Blood Cells; b) Cryoprecipitate AHF.
               Recall #B-1286/1287-7.
  CODE         Unit 21045-1515.
  MANUFACTURER United Blood Services, Fargo, North Dakota.
  RECALLED BY  Blood Systems Inc., Scottsdale, Arizona, by
                 letter dated July 25, 1997.  Firm-initiated
                 recall complete.
  DISTRIBUTION North Dakota, Louisiana.
  QUANTITY     1 unit of each component was distributed.
                               -6-REASON         Blood products, which tested non-reactive for
                 the                    antibody to the Hepatitis B core antigen
                 (anti-HBc), but             were collected from a donor
                 who was previously               repeatedly reactive for
                 anti-HBc.
  
  _______________
  PRODUCT      Recovered Plasma, Frozen.  Recall #B-1461-7.
  CODE         25FS25282, 25FS25283, 25FS25284, 25FS25286,
                 25FS25287, 25FS25285, 25FS25288, 25FS25289,
                 25FS25291, 25FS25292, 25FS25293, 25FS25294,
                 25FS25295, 25FS25296, 25FS25297, 25FS25298,
                 25FS25299, 25FS25300, 25FS25301, 25FS25302,
                 25FS25303, 25FS25304, 25FS25306, 25FS25307,
                 25FS25308, 25FS25309, 25FS25310.
  MANUFACTURER American Red Cross Blood Services, Syracuse,
                 New York.
  RECALLED BY  Manufacturer, by E-mail and by fax on June 28,
                 1996.  Firm-initiated recall complete.
  DISTRIBUTION California and Switzerland.
  QUANTITY     27 units were distributed.
  REASON       Blood product was prepared from Whole Blood
                 which may have reached unacceptable
                 temperature during shipment.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1477-7.
  CODE         Unit #10239-8348.
  MANUFACTURER United Blood Services, Scottsdale, Arizona.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 telephone on September 13, 1994.  Firm-initiated recall complete.
  DISTRIBUTION Arizona.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was labeled with an extended
                 expiration date.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1478-7.
  CODE         Unit #10524-3527.
  MANUFACTURER United Blood Services, Scottsdale, Arizona.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 telephone on February 28, 1997.  Firm-initiated recall complete.
  DISTRIBUTION Arizona.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was prepared from a unit of
                 Whole                  Blood with an extended collection time
                 and designated as                a "difficult collection."
   
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1479-7.
                               -7-CODE           Unit numbers: 10521-6637, 10521-6636, 
               10521-6635.
  MANUFACTURER United Blood Services, Scottsdale, Arizona
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 telephone on January 22, 1997.  Firm-initiated
                 recall complete.
  DISTRIBUTION Arizona.
  QUANTITY     3 units were distributed.
  REASON       Blood products were labeled with extended
                 expiration dates.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1480-7.
  CODE         Unit numbers: 10238-3924, 10238-3920,
                 10238-3926, 10238-2741, 10238-2737,
                 10238-2739, 10238-4953, 10238-3928,
                 10238-3929, 10238-4792.
  MANUFACTURER United Blood Services, Scottsdale, Arizona.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 letter dated August 22, 1994.  Firm-initiated
                 recall complete.
  DISTRIBUTION Arizona.
  QUANTITY     10 units were distributed.
  REASON       Blood products were shipped at an unacceptable
                 temperature. 
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1495-7.
  CODE         Unit numbers: 36247-0348, 36247-0134, 
               36247-0132.
  MANUFACTURER United Blood Services, Ventura, California.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 telephone on October 1, 1996.  Firm-initiated
                 recall complete.
  DISTRIBUTION California.
  QUANTITY     3 units were distributed.
  REASON       Blood products may have been stored at an
                 unacceptable temperature.
  
  
  RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS I ==========
  _______________
  PRODUCT      Solar 8000 Bedside Patient Monitor.  
               Recall #Z-908-7. 
  CODE         Serial numbers beginning with B7, C7, D7, E7,
                 FAll monitors manufactured between February 1,
                 1997 and July 10, 1997, with s7 or G7.
  MANUFACTURER Marquette Electronics, Inc., Milwaukee,
                 Wisconsin.
  RECALLED BY  Manufacturer, by letter on August 15, 1997. 
                 Firm-initiated field correction ongoing.
  DISTRIBUTION Nationwide and international.. 
  QUANTITY     1,769 units were distributed.
                               -8-REASON                        The speaker of the monitor may fail to audibly
                                alarm due to the inappropriate length of the
                                speaker lead wire.
  
  _______________
  PRODUCT      Lead Aprons, Collars, and Gonadal Shields.
               Recall #Z-912/914-7.
  CODE         Products received from vendor between 5/97 and
                 7/97.  No lot number or catalog number.
  MANUFACTURER Burlington Medical, Inc., Newport News,
                 Virginia.
  RECALLED BY  Manufacturer, by letter on July 15, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Nationwide.
  QUANTITY     Undetermined.
  REASON       The radiation protection devices contain lead
                 contaminated with small amounts of radioactive
                 substances.
  
  _______________
  PRODUCT      Ultra-Flex Adult Anesthesia Breathing
                 Circuits:
               1.  Product No. 51213F-6154 
               2.  Product No. 5613B-61 
               3.  Product No. 5613F-61 
               4.  Product No. 5613F-80 
               5.  Product No. 5753B-61 
               6.  Product No. 5753F-10021 
               7.  Product No. 61453B-8015X 
               8.  Product No. 6753F-13621 
               9.  Product No. 8613B-164 
               10. Product No. 8613B-84X 
               11. Product No. 8613F-13633 
               12. Product No. 8613F-164E 
               13. Product No. 8613F-61 
               14. Product No. 8613F-6121 
               15. Product No. 8613F-80 
               16. Product No. 8613F-84 
               17. Product No. 8613F-8433 
               18. Product No. 8752B-84X 
               19. Product No. 8753B-61 
               20. Product No. 8753F-104 
               21. Product No. 8753F-231 
               22. Product No. 8753F-248 
               23. Product No. 8753F-249 
               24. Product No. 8753F-61 
               25. Product No. 90337 
               26. Product No. 90358 
               27. Product No. Q5753F-28B31.
               Recall #Z-915/941-7. 
  CODE         Lot numbers:
               1. 1947E7
               2. 0236F7, 1948E7
                               -9-               3. 0237F7, 1337E7
               4. 1238F7
               5. 0101F7
               6. 2296E7
               7. 0680F7
               8. 2058E7
               9. 0586F7, 1341E7
               10. 1139E7, 0243F7, 0213E7, 1205F7
               11. 0101F7, 2300E7, 1728E7
               12. 1342E7
               13. 0244F7, 1952E7, 2158E7
               14. 0587F7
               15. 1037F7, 1343E7, 1953E7
               16. 1282F7
               17. 1212E7
               18. 0504F7
               19. 0505F7, 1603E7
               20. 2061E7, 0104F7
               21. 0105F7
               22. 0247F7
               23. 2062E7, 0106F7, 1346E7, 0248F7
               24. 0250F7, 2041E7
               25. 1734E7
               26. 2042E7
               27. 0002F7.
  MANUFACTURER King Systems Corporation, Noblesville,
                 Indiana.
  RECALLED BY  Manufacturer, by fax on August 11, 1997. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and Singapore.
  QUANTITY     328 cases were distributed.
  REASON       The expandable tube is breaking around the
                 circumference of the tube.
  
  
  SAFETY ALERTS: ==============================================
  _______________
  PRODUCT      Autopheresis-C Plasmapheresis System.
               Safety Alert #N-028-7.
  CODE         Model A-200, Model A-201, Model A-401.
  MANUFACTURER Baxter Healthcare Corporation, Largo, Florida.
  ALERTED BY   Baxter Healthcare Corporation, Deerfield,
                 Illinois, by letter on August 4, 1997.
  DISTRIBUTION Nationwide.
  QUANTITY     5,229 instruments were distributed.
  REASON       The use of 250 ml anticoagulant containers may
                 result in the potential for donor air embolus
                 when performing plasmapheresis with the
                 Autopheresis-C instrument.
  
                             -10-
  
  END OF ENFORCEMENT REPORT FOR OCTOBER 1, 1997.  BLANK PAGES MAY FOLLOW.
                                   ####