  PRODUCT      Surimi (Imitation Crab) Spread.  The 8 oz. 
               containers are labeled as "Surimi Crab Spread" 
               and the 4 lb. containers are labeled as
               "Imitation Crab Spread".  Recall #F-819-7.
  CODE         8 oz. containers with "Use by" date "8/31/97";
               4 lb. containers
               are uncoded and thus are all under recall.
  MANUFACTURER Silver Creek Farms, Twin Falls, Idaho.
  RECALLED BY  Manufacturer, on or about August 12, 1997, 
               and by press release on August 13, 1997.           
               Firm-initiated recall complete.
  DISTRIBUTION Oregon, Utah, Idaho, Wyoming, Nevada.
  QUANTITY     558 pounds distributed.  This is broken down 
               into 53 cases (12-8 ounce cups per case) and 20    
               cases (3-4 pound tubs per case).  
  REASON       The product is contaminated with Listeria
               monocytogenes.
  
  _______________
  PRODUCT      Strawberry Frozen Yogurt, in 3 gallon tubs, 1/2
               gallon, quart, and pint containers, under the following
               labels:  Our Own, Licks, and Muhlenberg Foods.
               Recall #F-820-7.
  CODE         161.
               MANUFACTURER Muhlenberg Foods, Reading,
               Pennsylvania.RECALLED BY Manufacturer, by telephone 
               on July 31, 1997, and by press release 
               on August 1, 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Pennsylvania.
  QUANTITY     12 pint, 40 quart, 38 half gallon, and 11 3-gallon
               containers.
  REASON       Product is contaminated with Listeria
               monocytogenes.
  
  _______________
  UPDATE       Refrigerated potato salad products, which appeared
               in the September 3, 1997 Enforcement Report is being
               updated as follows:
  PRODUCT      Refrigerated Potato Salad Products produced by
               Orval Kent Food Products in 1 lb., 2 lb. and 3 lb.
               plastic retail containers under the Orval Kent,
               Pathmark, Demoulas, Giant of Landover Maryland,
               Acme Markets and Marsh Supermarkets labels and
               five and ten lb. bulk containers under the
               Orval Kent (also 30 lb.), Salad Plus, Signature Salads,
               Signature Delights, Sysco Classic, Sysco Block
               & Barrel, Alliant, PYA Monarch, Matthew's Salad
               House and Metromedia labels as follows:
               a) Orval Kent Deluxe Potato Salad, 3/10 lb., UPC
               79453-13005
               b) Alliant Prestige Mustard Potato Salad with
               Salad Dressing,
               Alliant Foodservice Inc., Dist., Deerfield, IL 
               60015, 3/10 lb.,
               UPC 08077-84131
               c) Alliant Prestige American Potato Salad with
               Salad Dressing,
               3/10  lb., UPC 08077-84132
               d) Alliant Prestige American Potato Salad with
               Mayonnaise, 3/10 lb.,  UPC 08077-84138 
               e) Sliced Mustard Potato Salad 
                i) PYA Monarch, Dist. by P.Y.A./Monarch, 
                Virginia Beach, VA 23455-3701, 1/10
                lb., UPC 86270-01457
                ii) Orval Kent, 3/10 lb., UPC 79453-40204
               f) Diced Potato Salad with Egg
                i) Super G, Distributed by Super G, Inc.,
                Landover, MD 20785 12/1 lb.
                ii) Orval Kent, 3/10 lb., UPC 79453-40304
               g) Orval Kent Classic Potato Salad, 2/5 lb., UPC
                79453-40756
               h) Sliced Potato Salad with Egg
               i) Acme Corner Deli Cuisine, Distributed by
                   Acme Markets, Inc., Malvern, PA 19355,
                   12/1 lb., UPC 79453-04453
               ii) Sysco, Distributed by Sysco Corporation,
               Houston, TX 77077-2099, 3/10 lb., UPC
               74865-10108
                                                                 
              iii) Orval Kent, 3/10 lb., UPC 79453-41004
                  1/10 lb., UPC 79453-44106
               i) Baked Style Potato Salad
               i) Orval Kent, 2/5 lb., UPC 79453-41756
               ii) Alliant Prestige, 2/5 lb., UPC
               08077-70502
               j) Orval Kent Amish Potato Salad, 1/10 lb.,
               UPC 79453-41804
               k) Metromedia Steakhouses Company, L.P. Southern
                 Style Potato
                 Salad, Mfg. by Orval Kent Food Company, Inc.,                                                               
                 Wheeling, IL
                 60090 for Exclusive Use in the Metromedia
                 Steakhouses Company,
                 LP, 3/10 lb.
               l) German Potato Salad
               i) Sysco, 1/10 lb., UPC 74865-07978
               ii) Pathmark, Packed for Pathmark Stores,
               Inc., Woodbridge, NJ 07086, 12/1 lb.,
               UPC 020670
                iii) Orval Kent, 2/5 lb., UPC 79453-45200
                                   1/10 lb., UPC 79453-45204
               m) Diced Potato Salad
                I) Orval Kent, 6/5 lb., UPC 79453-01313
                   6/3 lb., UPC 79453-42815
                   3/10 lb., UPC 79453-42883
                                   6/2 lb., UPC 79453-42885
                                   1/10 lb., UPC 79453-42886
                                   12/1 lb., UPC 79453-42888
                ii) Sysco, 3/10 lb., UPC 74865-18777
                iii) Orval Kent Canadian label, 1/10 lb., UPC
                79453-20842 6/3 lb., UPC 79453-28526
                n) Orval Kent 100% Fat Free Potato Salad, 1/10
                 lb.,  UPC 79453-43004
                o) Fat Free Roasted Garlic Red Potato Salad
                 i) Orval Kent Signature Delights, 2/5 lb.,
                 UPC 79453-43387
                 ii) Sysco Imperial, 2/5 lb., UPC 74865-40707
                 p) Sliced White Potato Salad
                 i) Demoulas, Distributed by Demoulas
                 Supermarket, Tewksbury, MA 01876, 12/1
                 lb., UPC 70226-65006
                 ii) Sysco, 3/10 lb.
                 iii) Orval Kent, 3/10 lb., UPC 79453-44001
                 and 1/10 lb.
                q) Sysco Reduced Fat Country Style Potato
                 Salad with Egg, 3/10 lb., UPC 74865-31311
                 r) Red Skinned Potato Salad
                 i) Orval Kent Signature Delights, 2/5 lb.,
                 UPC 79453-44787
                 ii) Alliant Prestige, 2/5 lb., UPC
                 08077-70503
                 s) Country Style Potato Salad with Egg
                 i) PYA Monarch, 1/10 lb., UPC 86270-07179
                 ii) Sysco, 3/10 lb., UPC 74865-18775
                   
                 iii) Orval Kent, 1/10 lb., UPC 79453-45904
                      1/30 lb., UPC 79453-45949
                  iv) Orval Kent Homestyle Egg & Potato
                   Salad, 2/5 lb., UPC 79453-45000
                  t) Southern Style Mustard Potato Salad
                  i) PYA Monarch, 1/10 lb., UPC 86270-01456
                  ii) Orval Kent, 1/10 lb. and 1/10 lb., UPC
                  79453-43104
                  u) Orval Kent Low Country Potato Salad, 1/10
                  lb.,  UPC 79453-45704
                  v) Sliced Potato Salad with Vegetables
                  i) Chilled Salad, Distributed by G.D.
                  Matthews & Sons, Charlestown, MA 02129,
                    3/10 lb.
                 ii) Pathmark, 12/1 lb. 6/2 lb. 6/3 lb.
                iii) Orval Kent, 6/3 lb., UPC 79453-45815
                     12/1 lb., UPC 79453-45888
                     3/10 lb., UPC 79453-45849
                w) Diced Mustard Potato Salad
                 i) Acme Corner Deli Cuisine, 12/1 lb., UPC
                 79453-05653 6/2 lb., UPC 79453-05953
                 ii) Orval Kent, 3/10 lb., UPC 79453-46301
                     6/3 lb., UPC 79453-46315
               x) Homestyle Potato Salad
                 i) Deli Fresh at Marsh, Dist. by Marsh
                   Supermarkets, Inc., Indianapolis, IND
                     46256-3350, 12/1 lb., UPC 79453-46425 
                 ii) Orval Kent, 1/10 lb., UPC 79453-46404
               y) Orval Kent Red Royale Potato Salad, 2/5
                  lb., UPC 79453-41456
               z) Sysco Southern Style Mustard Potato Salad,      
                    
                   3/10 lb.,  UPC 74865-33830
               aa) Super G Diced Potato Salad with
                    Vegetables, 6/2 lb., 3/10 lb., UPC 
                    79453-14001, 12/1 lb.
               bb) Alliant Prestige Gourmet Potato Salad, 2/5
                    lb.,  UPC 08077-70504
               cc) Orval Kent Signature Specialties Natural
                   Potato & Herb Salad 2/5 lb., UPC
                   79453-80056
               dd) Orval Kent Signature Specialties Red
                   Potato & Egg Salad Supreme, 2/5 lb., UPC
                   79453-80656
               ee) Sour Cream & Dill Potato Salad
                   i) Salads Plus, Dist by Orval Kent, 3/10
                      lb. and 1/10 lb., UPC 79453-88005
                  ii) Block & Barrel Deluxe Potato Salad,
                      Dist. by Sysco Corp., 2/5 lb., UPC
                      74865-40685.  Recall# F-684/F-713-7.
  CODE         All potato salad products with expiration dates
               prior to September 15, marked with the code "W7", 
               i.e. SEP15W7.
  MANUFACTURER Orval Kent Food Company, Inc., Wheeling,
               Illinois.
  RECALLED BY  Manufacturer beginning on August 4, 1997. 
               Firm-initiated recall
                 ongoing.
  DISTRIBUTION Nationwide, Puerto Rico, and Canada.
  QUANTITY     5,644,170 pounds were distributed.
  REASON       Potato salads may be contaminated with Listeria
               monocytogenes.
  
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
  _______________
  PRODUCT      Pepperidge Farm Frozen Assorted Mini & Large
               Danish Pastries:
               a) Pepperidge Farm Frozen Assorted Mini Danish
                  Pastries UPC# 510006747;
               b) Pepperidge Farm Frozen Assorted Large Danish
                 Pastries UPC# 510007812.  Recall #F-809/810-7.
  CODE         All product with "use by" dates between April 21,
               1997 and August 14, 1997.
  MANUFACTURER Campbell Soup National HQ, Camden, New Jersey.
  RECALLED BY  Pepperidge Farms, Inc., Norwalk, Connecticut, by
               letter dated February 20, 1997.
               Firm-initiated recall complete.
  DISTRIBUTION Nationwide.
  QUANTITY     a) 37,064 cases (96 pastries per case); b) 14,730
               cases (48 pastries per case) were distributed.
  REASON       The pastries contain excessive levels of sulfites.
  
  _______________
  PRODUCT      Cock brand Tapioca Pearl.  Recall #F-811-7.
  CODE         All including EMB 75113N (Best Before 1996).
  MANUFACTURER Unknown in Thailand.
  RECALLED BY  Hong Thai Foods Corporation, Brooklyn, New York,
               by letter dated
               April 7, 1997.  Firm-initiated recall complete.
  DISTRIBUTION New York, New Jersey, Pennsylvania, Connecticut,
               Massachusetts, Illinois, Virginia, Florida.
  QUANTITY     49 cases were distributed.
  REASON       Product contains the following undeclared colors: 
               FD&C Yellow, No. 5, FD&C Blue No. 1, and FD&C Red No. 3.
  
  _______________
  UPDATE       Brown Sugar, in 1, 2, 4, 25, 50 and 100 lb
               packages.  Recall #F-806-7, which appeared in the 
               September 10, 1997 Enforcement Report should read:
  CODE         Lot #46092/46106.
  RECALLED BY  Manufacturer, on or about May 5, 1996.
                              
  QUANTITY       Approximately 2 million pounds.
  
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
  _______________
  PRODUCT      Ensure Plus, vanilla flavored, high-calorie
               nutritional drink, packed in 8 fluid ounce plastic cans.  Recall #F-808-7.
  CODE         Cans are identified with lot code: 28778L5 EXP1
               MAY 98.
  MANUFACTURER Ross Products Division, Columbus, Ohio.
  RECALLED BY  Manufacturer, by telephone on June 24, 1997,
               followed by fax. 
               Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     2,208 cases were distributed.
  REASON       Product bears the label of another food and               
               contains erroneous label information regarding identity and nutrient
               content.
  
  
  RECALLS AND FIELD CORRECTIONS:  COSMETICS -- CLASS II =======
  _______________
  PRODUCT      Nectar Watermist a non alcohol body spray made
               from natural distilled water extracted from blossoms of
               orange trees packaged into .5 ounce, 4 ounce, 8 ounce
               and 16 ounce glass bottles.  Recall #F-807-7.
  CODE         The codes under recall consists of all product
               codes shipped prior to July 28, 1997 and includes the
               following items: 
               4 oz. Nectar WaterMist Item #FG-8120D-00           
               Packaging Lot #s 71611, 71621, 71631;
               8 oz. Nectar WaterMist, Item #FG-8123D-00,         
               Packaging Lot #s 71641, 71651; 
               8 oz. Nectar WaterMist (tester), Item #               
               PG-7427D-00, Packaging
               Lot #s 71641, 71651;
               Bulk lot #s
               .5 oz Nectar (tester) sample, Item # PG-7421D-00,
               Packaging lot #7F09A (only given to sales reps);
               Dramming/Nectar bulk (3 X 16 oz.), Item #PG-7740D-00,
               Packaging Lot#  7F09A.
  MANUFACTURER Allure Cosmetics, San Leandro, California  Herba
               Aromatica, Hayward, California (importer of scented  
               floral water).
  RECALLED BY  The Donna Karan Beauty Company (DKBC) New York,
               New York, by telephone visit and fax beginning on
               July 28, 1997, followed by letter on August 1, 1997.
               Firm-initiated recall ongoing.
  
 DISTRIBUTION  California, New York, Washington state, Oregon,
               Virginia, Pennsylvania, Maryland, New Jersey,
               Massachusetts, Connecticut, Florida, Minnesota, 
               Illinois, Wisconsin, Indiana, Michigan, Texas.
 QUANTITY      Item #FG-8120D-00 - 3,606 units - 4 oz Nectar
               Item #FG-8123D-00 - 1,967 units - 8 oz Nectar
               Item #PG-7427D-00 - 713 units - 8 oz Watermist
               Item #PG-7426D-00 - 4,028 units - 5 oz Nectar
               Item #PG-7440D-00 - 65 units - Dramming bulk 
               (3 x16 oz) were distributed. 
 REASON        Product is contaminated with Burkholderia cepacia,
               a pathogen.
  
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
  ===========
  _______________
  PRODUCT      Thiopental Sodium for Injection, USP, 5 gram/vial,
               in 200 ml vials, in 10 packs of vials of Thiopental
               Sodium without kits under the Gensia label, kits and
               vial ten packs) and the Ohmeda label (kits), for
               preparation of intravenous stock solution only.  
               Recall #D-265-7.
  CODE         Lot numbers:  P6P825,P7A817, P7A818 (Ohmeda Kit
               label),
               P6A317P1, P6P312, (Gensia label/10 pak) and
               P6S815 (Gensia Kit label).
  MANUFACTURER Gensia Laboratories, Ltd., Irvine, California.
  RECALLED BY  Manufacturer, by letter on April 21, 1997. 
               Firm-initiated recall complete.
  DISTRIBUTION Nationwide and Puerto Rico.
  QUANTITY     Approximately 7,900 kits and 9,000 vials in ten
               packs were distributed.
  REASON       Vials were found to contain glass particles.
  
  _______________
  PRODUCT      Captopril (Capoten) 25 mg Tablets, individually (1
               tablet) labeled, single blister packed tablets with
               labeled single unit plastic overwrap, Rx, used in the 
               treatment of hypertension.
               Recall #D-266-7.
  CODE         Lot #J5J65A EXP 8/99.
  MANUFACTURER ER Squibb & Sons, Inc., Division of Bristol Myers
               Squibb, Medical Department, Princeton, New Jersey.
  RECALLED BY  AmeriSource Health Services, doing business as
               American Health Packaging (AHP), Columbus, Ohio,
               by telephone on or about October 22, 1996.
               Firm-initiated recall ongoing.
  DISTRIBUTION Ohio, Tennessee, Massachusetts.
  
  QUANTITY     130 amber bags (13,000 tablets) were distributed. 
               Firm estimates none remains on the market.
  REASON       Labeling error:  Incorrect product identification
               - Plastic overwrap declares the product as Warfarin
               Sodium (Coumadin) 2 mg Tablets.  Each individual
               blister label is correctly labeled.
  
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
  _______________
  PRODUCT      Ciprofloxacin HCl (Cipro), 500 mg tablets,
               individually (1 tablet) labeled single blister packed
               tablets with a plastic overwrap, an antibiotic. 
               Recall #D-264-7.
  CODE         Lot #6BCJ EXP 2/99.
  MANUFACTURER Bayer Corporation Pharmaceutical Division, West
               Haven, Connecticut..
  RECALLED BY  AmeriSource Health Services, Columbus, Ohio, doing
               business as American Health Packaging (AHP), by
               telephone on or about March 11, 1997.  Firm-initiated
               recall ongoing.
  DISTRIBUTION Ohio, Tennessee, Massachusetts.
  QUANTITY     90 amber bags (9,000 tablets) were distributed;
               firm estimates none remains on the market.
  REASON       Labeling:  error on declared strength.  Plastic
               overwrap declares the product as 250 mg.
  
  _______________
  UPDATE       Recall #D-171-7, Regular Strength Enteric Coated
               Aspirin Tablets, 325 mg, packaged in 100, 250, and
               500 count bottles of various lots, which appeared in
               the May 14, 1997 Enforcement Report as a Class II 
               recall has been reclassified to a Class III recall.
  
  _______________
  UPDATE       Recall #D-259/260-7, Baclofen Tablets which
               appeared in the September 10, 1997 Enforcement Report
               should read:  MANUFACTURER: Biocraft Laboratories,
               Elmwood Park, New Jersey.
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1193-7.
  CODE         Unit #M46536.
  MANUFACTURER Lane Memorial Blood Bank, Eugene, Oregon.
  RECALLED BY  Manufacturer, by visit on October 8, 1996. 
               Firm-initiated recall complete.
  DISTRIBUTION   Oregon.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
               engaged in high risk behavior.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Red Blood Cells, Leukocytes
                  Reduced; c) Platelets; d) Fresh Frozen Plasma.
                  Recall #B-1344/1347-7.
  CODE         Unit numbers:  a) 29123-1232, 29092-0971,
               29128-0189, 29093-9146, 29119-4436, 29125-8658,
               29122-3879, 29082-2745, 29081-2570; b) 29131-8807;
               c) 29123-1232, 29092-0971, 29093-9146, 29125-8658,
               29122-3879, 29131-8807; d) 29123-1232, 29092-0971,     
               29128-0189, 29122-3879, 29082-2745, 29081-2570.
  MANUFACTURER United Blood Services, Chicago, Illinois.
  RECALLED BY  Manufacturer, by letter on either May 30, 1996,
               June 5, 1996, or July 5, 1996.  Firm-initiated recall
               complete.
  DISTRIBUTION Illinois, Florida, Switzerland.
  QUANTITY     a) 9 units; b) 1 unit; c) 6 units; d) 6 units were
               distributed.
  REASON       Blood products tested negative for the antibody to
               the human T-lymphotropic virus I (anti-HTLV-1) but
               were collected from donors who previously tested
               repeatedly reactive for anti-HTLV-1, Western blot
               negative or indeterminate.
  
  _______________
  PRODUCT      Platelets, Pheresis.  Recall #B-1350-7.
  CODE         Unit #21P69503.
  MANUFACTURER American Red Cross, Portland, Oregon.
  RECALLED BY  Manufacturer, by telephone on June 12, 1997, and
               by letter dated June 17, 1997. 
               Firm-initiated recall complete.
  DISTRIBUTION Oregon.
  QUANTITY     1 unit was distributed.
  REASON       Blood product had an unacceptable platelet count.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1368-7.
  CODE         Unit #25FR70821.
  MANUFACTURER American Red Cross Blood Services, Syracuse, New
               York.
  RECALLED BY  Manufacturer, by letter dated November 3, 1995. 
               Firm-initiated recall complete.
  DISTRIBUTION New York.
  QUANTITY     1 unit was distributed.
  REASON       Platelets were prepared from a unit of whole blood
               designated as "difficult" phlebotomy.
  
  _______________
  PRODUCT      Red Blood Cells, Washed.  Recall #B-1378-7.
  CODE         Unit numbers:  16337-4457, 16337-4456, 
               16337-4474, 16337-7805.
  MANUFACTURER Blood Systems, Inc., Hattiesburg, Mississippi.
  RECALLED BY  Manufacturer, by letter dated March 21, 1997. 
               Firm-initiated recall ongoing.
  DISTRIBUTION Mississippi.
  QUANTITY     4 units were distributed.
  REASON       Blood products were labeled with extended
               expiration dates.
  
  _______________
  PRODUCT      a) Cryoprecipitate; b) Fresh Frozen Plasma. 
               Recall #B-1387/1388-7.
  CODE         Unit numbers:  a) 8580139 and 8551475; 
               b) 8574398 and 8526946.
  MANUFACTURER The Blood Center, New Orleans, Louisiana.
  RECALLED BY  Manufacturer, by letter and or telephone between
               September 8, 1995 and October 9, 1995.
               Firm-initiated recall complete.
  DISTRIBUTION Louisiana.
  QUANTITY     2 units of each component were distributed.
  REASON       Blood products contained anti-D antibodies, but
               were labeled as negative for unexpected antibodies.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets.  
               Recall #B-1402/1403-7.
  CODE         Unit numbers:  a) 16339-7346, 16339-7344,
               16339-7341; b) 16339-7347, 16339-7346, 16339-7344,
               16339-7343, 16339-7342, 16339-7341, 16339-7340,
               16339-7000.
  MANUFACTURER Blood Systems, Inc., Meridian, Mississippi.
  RECALLED BY  Manufacturer, by telephone on June 3, 1997. 
               Firm-initiated recall
               complete.
  DISTRIBUTION Mississippi.
  QUANTITY     a) 3 units; b) 8 units were distributed.
  REASON       Blood products were tested using samples that were
               stored at an unacceptable temperature.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1404-7.
  CODE         Unit numbers:  16339-8033, 16339-8036, 16339-8037,
               16339-8039, 16339-8041, 16339-8042, 16339-8044,
               16339-8045, 16339-8046, 16339-8047, 16339-8048,
               16339-8050, 16339-8051, 16339-8052.
  MANUFACTURER Blood Systems, Inc., Meridian, Mississippi.
  RECALLED BY  Manufacturer, by telephone on June 9, 1997. 
               Firm-initiated recall complete.
  DISTRIBUTION Mississippi.
  QUANTITY     14 units were distributed.
  REASON       Blood products were tested using samples that were
               stored at an unacceptable temperature.
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1408-7.
  CODE         Unit #8637621.
  MANUFACTURER The Blood Center, New Orleans, Louisiana.
  RECALLED BY  Manufacturer, by telephone June 7, 1996. 
               Firm-initiated recall complete.
  DISTRIBUTION Louisiana.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was not properly tested for HIV-Ag.
  _______________
  UPDATE       United Blood Services, Scottsdale, Arizona recall
               which appeared in the September 3, 1997 Enforcement
               Report should read:
               PRODUCT:  a) Red Blood Cells; b) Fresh Frozen
               Plasma; c) Recovered Plasma.  
               Recall #B-1281/1283-7.
               CODE:   Unit numbers:  10230-8872, 10175-4045.
               QUANTITY:  4 units were distributed.
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
  _______________
  PRODUCT      Recovered Plasma. Recall #B-1348-7.
  CODE         29093-9146, 29119-4436, 29125-8658, 
               29131-8807.
  MANUFACTURER United Blood Services, Chicago, Illinois.
  RECALLED BY  Manufacturer, by letter on either May 30, 1996,
               June 5, 1996 or July 5, 1996.  Firm-initiated recall
               complete.
  DISTRIBUTION Illinois, Florida, Switzerland.
  QUANTITY     4 units were distributed.
  REASON       Blood products tested negative for the antibody to
               the human T-lymphotropic virus (anti-HTLV-1), but
               were collected from donors who previously tested
               repeatedly reactive for anti-HTLV-1, Western blot
               negative or indeterminate.
  _______________
  PRODUCT      Fresh Frozen Plasma.  Recall #B-1385-7.
  CODE         Unit #16335-3005 and 16335-3007.
  MANUFACTURER Blood System, Inc., doing business as United Blood
               Services,
               Hattiesburg, Mississippi.
  
  RECALLED BY  Manufacturer, by telephone on October 19, 1996 and
               by letter on November 22, 1996.  Firm-initiated recall
               complete.
  DISTRIBUTION Mississippi.
  QUANTITY     2 units were distributed.
  REASON       Fresh Frozen Plasma was not frozen within eight
               hours of phlebotomy.
  
  _______________
  PRODUCT      Red Blood Cells, Deglycerolized.  
               Recall #B-1386-7.
  CODE         Unit numbers:  29444-5147, 29444-5055.
  MANUFACTURER United Blood Services, Chicago, Illinois.
  RECALLED BY  Manufacturer, by telephone on June 23, 1996, and
               by letter on July
                 30, 1996.  Firm-initiated recall complete.
  DISTRIBUTION Illinois.
  QUANTITY     2 units were distributed.
  REASON       Deglycerolized Red Blood Cells were not treated
               with rejuvenating solution prior to freezing.
  
  
  RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS
II
  =========
  _______________
  PRODUCT      Lead Protective Aprons used to shield patients
               during x-ray procedures.  Recall #Z-904-7.
  CODE         Serial numbers:  6502-6523, 6532-6537, 6551,
               6555-6559, 6567-6569, 6578-6579, 6585-6642,
               6648-6668, 6670-6678, 6681-6686, 1237-1238, 3896-3909,
               3911-1912, 3914-3919, 3966-3968, 3975-3976, 3993,
               3998, 4040-4042, 1237-1238.
  MANUFACTURER Pulse Medical, Inc., Davie, Florida.
  RECALLED BY  Manufacturer, by letter on June 10 and 16, 1997. 
               Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and Canada.
  QUANTITY     153 aprons were distributed.
  REASON       Radiation protection devices contain lead contaminated 
               with small amounts of radioactive substances.
  
  _______________
  PRODUCT      Abiomed BVS 5000 Bi-Ventricular Support System
               Transistor on Compressor Driver.  The device
               (PMA #900023) provides ventricular assist and 
               restores hemodynamics in patients suffering
               severe cardiac dysfunction.  It consists of         
               three major components:  the BVS Console,
               the BVS pumps and the BVS Cannulae.  
               Recall #Z-909-7.
  CODE         Console Catalog No. 0005-000, Console serial        
               numbers:  1314,1336, 1339, 1341-1352, 1352, 1355.
  MANUFACTURER Abiomed, Inc., Danvers, Massachusetts.
  RECALLED BY  Manufacturer, by letter dated June 30, 1997.         
               Firm-initiated recall complete.
  DISTRIBUTION Arkansas, California, Florida, Illinois, Louisiana,
               Michigan, Missouri, New Jersey, Oklahoma, Texas.
  QUANTITY     17 consoles were distributed.
REASON         The console can stop during use and go into          
               continuous alarm.
  
                             
  
  END OF ENFORCEMENT REPORT FOR SEPTEMBER 17, 1993.  BLANK PAGES