FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.


August 13, 1997                                       97-33

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT   Progresso 99% Fat Free White Cheddar Potato Soup,"
          distributed to Von's Markets in California and Nevada. 
          Recall #F-626-7.
CODE K6N14.
MANUFACTURER   Progresso Quality Foods, Vineland, New Jersey.
RECALLED BY    The Pillsbury Company, Minneapolis, Minnesota, by
               letter on July 1, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   California and Nevada.
QUANTITY  240 cases (12 cans per case) were distributed.
REASON    The product contains undeclared clams.  Some cans
          labeled as 99% Fat Free White Cheddar Potato Soup
          actually contained 99% Fat Free New England Clam
          Chowder.

_______________
PRODUCT   Wakim's Homestyle Hommus with Roasted Peppers, in 8
          ounce plastic tubs.  Recall #F-627-7. 
CODE June 29.
MANUFACTURER   Wakim's Foods Inc., Bristol, Pennsylvania.
RECALLED BY    Manufacturer, by press release and telephone on
               July 3, 1997, followed by visit on July 4 and 5,
               1997.  Completed recall resulted from sample
               analysis and followup by New York Department of
               Agriculture and Markets.
DISTRIBUTION   Pennsylvania, New Jersey, New York.
QUANTITY  Firm estimates none remains on the market.
REASON    The product was contaminated with Listeria
          monocytogenes.

_______________
PRODUCT   7" Carrot Cakes, sold in a clear plastic dome
          container.  Recall #F-628-7.
CODE The product is sold under various labels but can be
     identified by the last five digits of the UPC code "00722". 
     Only product purchased before July 7, 1997, is affected.
MANUFACTURER   Knaub's Cakes, Inc., York, Pennsylvania.
RECALLED BY    Manufacturer, by telephone on July 3, 1997,
               followed by fax, and by press release on July 8,
               1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Georgia. 
QUANTITY  Undetermined.
REASON    The product contains undeclared walnuts.

_______________
PRODUCT   Chocolate ice cream in half gallon paper containers,
          under the Osborn label.
     Recall #F-630-7.
CODE All codes.
MANUFACTURER   Osborn Dairy, Sault Saint Marie, Michigan.
RECALLED BY    Manufacturer, by press release by the firm in
               cooperation with the Michigan Department of
               Agriculture on July 3, 1996, followed by radio and
               newspaper announcements, and visits.  Completed
               recall resulted from sample analysis and followup
               by the Michigan Department of Agriculture.
DISTRIBUTION   Michigan.
QUANTITY  Firm estimates none remains on the market.
REASON    The chocolate ice cream is contaminated with Listeria
          monocytogenes. 


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
_______________
PRODUCT   Creative Products' Candy: a) Fish Tasting Candy; b)
          Fart Candy, individually wrapped pieces of candy
          attached to a display card.
     Recall #F-620/621-7.
CODE All product in distribution.
MANUFACTURER   F.B. Washburn Candy Corporation, Brockton,
               Massachusetts.
RECALLED BY    Creative Products Mfg., Inc., Bath, Pennsylvania,
               by letter dated July 25, 1997.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.

                               -2-QUANTITY  a) 5,000 packages; b) 2,000 packages were distributed.
REASON    a) Product contains undeclared FD&C Yellow No. 5 and
          FD&C Blue No. 1.
     b) Product contains undeclared FD&C Yellow No. 5, FD&C
     Yellow No. 6, FD&C Blue No. 1, and FD&C Red No. 40.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
_______________
PRODUCT   Creative Products Hot Sweets Candy, individually
          wrapped pieces of candy attached to a display card. 
          Recall #F-622-7.
CODE All product in distribution.
MANUFACTURER   F.B. Washburn Candy Corporation, Brockton,
               Massachusetts.
RECALLED BY    Creative Products Mfg., Inc., Bath, Pennsylvania,
               by letter dated July 25, 1997.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY  100 packages were distributed.
REASON    Product contains undeclared FD&C Red No. 40.

_______________
PRODUCT   Hillside Light Sour Cream, in 16 ounce plastic cups. 
          Recall #F-623-7.
CODE 07-19.
MANUFACTURER   Hillside Dairy Products, Cleveland Heights, Ohio.
RECALLED BY    Manufacturer, by telephone on June 10, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY  6,117 containers were distributed.
REASON    The product is misbranded because the label identifies
          it as "Light Sour Cream" when it in fact is regular
          sour cream.

_______________
PRODUCT   Product: Bean Sauce and Bean Curd Products:
      a) Bean Sauce (with chili), Net Wt. 16 oz. (1 lb.) or 454
     grams, in glass jars, 24 jars per case, packed by Chan Moon
     Kee 7-10 Chung Ching Street  Hong Kong
     b) Bean Curd Preserve: Net Wt. 10 oz. (300 g) in glass jars,
     48 jars per case.  Sze Chuan Food Products Co., Ltd. Product
     of Taiwan 
     c) Liu Ma Kee Preserved Bean Curd: Net wt. 454 g oz. (1 lb.)
     in glass jars, 24 jars per case. Product of Hong Kong
     d) Pearl River Bridge Fu Chung Brand Bean Curd, 16 oz. (1
     lb.) in glass jars, 36 jars per case.  Recall #F-616/619-7.  
                               -3-CODE None.
MANUFACTURERS/ a) Chan Moon Kee, Hong Kong
PACKERS   b) Szechuan Food Products Co. Ltd., Taiwan ROC
     c) Chan Moon Kee, Hong Kong; d) China National Cereals,
     Oils, & Foodstuffs I/F Co., The People's Republic of China. 
RECALLED BY    Blooming Imports, Inc., Brooklyn, New York
               (importer), by letter on November 4, 1996. 
               Completed recall resulted from sample analysis and
               followup by the New York State Department of
               Agriculture and Markets.
DISTRIBUTION   New York and New Jersey.
QUANTITY  a) 28 cases (24 cans per case); b) Undetermined; c) 46
          cases (24 cans per case);
     d) 90 cartons (36 jars per case were distributed.
REASON    The products are held in leaking containers.  In
          addition a) - c) are low acid canned foods for which no
          processes have been filed.  Products a) - b) are also
          contaminated with filth.


RECALLS AND FIELD CORRECTIONS:  COSMETICS -- CLASS II =======
_______________
PRODUCT   a) Gentle Lotion Cleanser and b) Triple S Gentle Wash
          in 500 ml plastic pouches packed 18 per case.  Recall
          #F-624/625-7.
CODE a) Lot numbers: 7040351001, 7041161001,  7041231001; b)
     7040941001.
MANUFACTURER   Sani-Fresh International, San Antonio, Texas.
RECALLED BY    Manufacturer, by letter dated July 9, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY  a) 3,640 cases (18 per case); b) 554 cases (18 per
          case) were distributed.
REASON    The product is contaminated with Pseudomonas
          aeruginosa.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
===========
_______________
PRODUCT   Maximum Strength Ecotrin Caplets, 500 mg, in 60 count
          bottles, indicated as an analgesic/anti-inflammatory
          agent.
     Recall #D-239-7.
CODE Lot numbers:  6I02A and 5H14A.
MANUFACTURER   Smithkline Beecham Pharmaceutical Company, Cidra,
               Puerto Rico.
RECALLED BY    Smithkline Beecham Consumer Healthcare,
               Parsippany, New Jersey, by letter on March 10,
               1997.  Firm-initiated recall ongoing.

                               -4-DISTRIBUTION   Nationwide.
QUANTITY  1,493 cases of lot 6I02A and 1,633 cases of lot 5H14A
          (36 bottles per case) were distributed; firm estimated
          that 150 cases remained on market at time of recall
          initiation.
REASON    Product failed dissolution testing (stability).

_______________
PRODUCT   Health Safe Anti-Bacterial/Antiseptic Polymer Lotion
          *** Safe Effective Skin Protection Against Infection
          (Triclosan 0.3%), in 2, 4, 8, 16, and 32 fluid ounce
          bottles.
     Recall #D-242-7.
CODE Lot #"4".
MANUFACTURER   Stanford Personal Care Manufacturer, Saugus,
               California.
RECALLED BY    AmeriCare Health Products, Marion, Illinois
               (distributor), by telephone on July 1, 1997,
               followed by letter July 7, 1997.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Tennessee, Illinois, Virginia, South Carolina,
               Wisconsin, Indiana, Maryland, Pennsylvania.
QUANTITY  312 - 2 oz. bottles, 192 - 4 oz. bottles, 204 - 8 oz.
          bottles, 432 - 16 oz. bottles and 81 - 32 oz. bottles
          were distributed, with the firm estimating that 30% of
          the product remains on the market.
REASON    Product contamination with Pseudomonas Aeruginosa.

_______________
PRODUCT   Oxygen, USP in D and E high pressure cylinders for
          prescription use.  Recall #D-243-7.
CODE All lots.
MANUFACTURER   Prescription Oxygen Services, Sault Saint Marie,
               Michigan.
RECALLED BY    Manufacturer, by visit.  Firm-initiated recall
               completed on July 10, 1997.
DISTRIBUTION   Michigan.
QUANTITY  Firm estimates none remains on the market.
REASON    Current good manufacturing practice deviations.

_______________
PRODUCT   Univasc Tablets, Moexipril Hydrochloride), 7.5 mg, in
          bottles of 100, used in the treatment of hypertension. 
               Recall #D-244-7.
CODE Lot 31411.
MANUFACTURER   Schwarz Pharma AG, Monheim, Germany.

                               -5-RECALLED BY    Schwarz Pharma, Inc., Mequon, Wisconsin, by letter
               on July 28, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY  13,010 bottles were distributed; firm estimated that
          less than 10 percent of product (1,301 bottles)
          remained on market at time of recall initiation.
REASON    Tablet mix-up:  15 mg tablet found in 7.5 mg bottle.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Amway FormuCare Hydrocortisone Cream USP, 1%, Net
               Wt. 28.35 g (1 oz.), in cartons labeled Amway
               FormuCare Triple Antibiotic Ointment,    
               BACITRACIN-NEOMYCIN SULFATE-POLYMYXIN B SULFATE
               OINTMENT.  Recall #D-240-7.
CODE Lot K500 EXP 7/00.
MANUFACTURER   Thames Pharmacal Company, Ronkonkoma, New York.
RECALLED BY    Amway Corporation, Ada, Michigan, by letter on
               April 25, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY  7,464 cases (4 units per case) were distributed.
REASON    Packaging mix-up:  Some correctly labeled tubes of
          hydrocortisone cream were packaged into exterior
          cartons labeled as triple antibiotic ointment.

_______________
PRODUCT   Tripelennamine Hydrochloride Tablets, USP, 50 mg, in
          bottles of 100 and 1,000, under the Schein label, an
          anti-histamine.  
     Recall #D-245-7. 
CODE Lot No. D6G1807. 
MANUFACTURER   Danbury Pharmacal, Inc., Danbury, Connecticut.
RECALLED BY    Manufacturer, by letter dated May 28, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY  448 1000-tablet bottles and 384 100-tablet bottles were
          distributed.
REASON    Subpotency due to low tablet weight.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT   Plasma (Cryo-Poor).      Recall #B-994-7.

                               -6-CODE Units: 29134-5599, 29134-5280, 23134-6966, 29133-9342,
     29133-9795, 29134-4614, 29134-6951,  29134-5281.
MANUFACTURER   United Blood Services,             Chicago,
               Illinois.
RECALLED BY    Manufacturer, by telephone on May 20, 1997, and by
               letter on July 17, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Illinois.
QUANTITY  8 units were distributed. 
REASON    Blood product was subject to quarantine because it may
          have been improperly tested for viral markers. 

_______________
PRODUCT   Red Blood Cells.  Recall #B-1077-7.
CODE Unit #30166-7350.
MANUFACTURER   United Blood Services, Bismarck, North Dakota.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               telephone on March 21, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   North Dakota.
QUANTITY  1 unit was distributed.
REASON    Blood product was prepared from a unit of whole blood
          which exceeded the whole blood to anticoagulant ratio.

_______________
PRODUCT   a) Red Blood Cells; b) Platelets;
     c) Fresh Frozen Plasma; d) Recovered Plasma.
     Recall #B-1134/1137-7.
CODE Unit numbers: a) 16047-5747, 16052-7307
     b) 16047-5747; c) 16047-5747
     d) 16052-7307.
MANUFACTURER   United Blood Services, Meridian, Mississippi.
RECALLED BY    Blood Systems Inc., Scottsdale, Arizona, by letter
               on March 13, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Mississippi, Florida, Switzerland.
QUANTITY  a) 2 units; b-d) 1 unit of each component was
          distributed.
REASON    Blood products were drawn from an unsuitable donor. 

_______________
PRODUCT   Red Blood Cells.  Recall #B-1142-7.
CODE Units  49FJ57215 and 49FJ57223.
MANUFACTURER   American Red Cross Blood Services, Tulsa,
               Oklahoma.  
RECALLED BY    Manufacturer, by telephone on June 6, 1997, and by
               letter dated June 9, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Texas and Oklahoma.
QUANTITY  2 units were distributed.
                               -7-REASON    Blood products were collected from donors whose body
          temperature and arm inspection had not been documented. 

_______________
PRODUCT   Red Blood Cells and Fresh Frozen Plasma.  Recall #B-1144/1145-7.
CODE Unit #10219-4509.
MANUFACTURER   United Blood Services, Scottsdale, Arizona. 
RECALLED BY    Blood Systems, Inc.,               Scottsdale,
               Arizona, by letter dated February 2, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Arizona.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products tested negative for viral markers,                but
          were collected from a donor who previously tested                repeatedly reactive for the
antibody to the human                immunodeficiency virus type 1 (anti-HIV-1), the
          antibody            to the human T-lymphotropic virus
          type I (anti-HTLV-I),              and the antibody to
          the hepatitis C virus encoded                antigen
          (anti-HCV). 

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; 
     c) Platelets, for further manufacture;
     d) Plasma; e) Recovered Plasma.
     Recall #B-1146/1150-7.
CODE Unit numbers:  a) 16J10119, 16G02869, 16K31428, 16H98369,
     16L75196, 16G07678, 16F85233, 16F85561; b) 16L75196,
     16G07678, 16H98369; c) 16K31428; d) 16F85233 and 16F85561;
     e) 16J10119, 16G02869, 16K31428, 16L75196, 16G07678,
     16H98369.
MANUFACTURER   American Red Cross Blood Services,                Columbus, Ohio.
RECALLED BY    Manufacturer, by letters dated December 6, 1995 or
               July 5, 1996, and on January 1, 1996.  Firm-initiated recall complete.
DISTRIBUTION   Ohio,  California, New York, Switzerland.
QUANTITY  a) 8 units; b) 3 units; c) 1 unit; d) 2 units; e) 6
          units were distributed.
REASON    Blood products tested repeatedly reactive for the                hepatitis B surface
antigen (HBsAg), and were subjected
                         to additional retesting.

_______________
PRODUCT   Platelets, Pheresis.  Recall #B-1170-7.
CODE Units FH7582A and FH7582B.
MANUFACTURER   Inova Blood Donor Services,             Annandale,
               Virginia.
RECALLED BY    Manufacturer, by letter dated September 9, 1996. 
               Firm-initiated recall complete.
                               -8-DISTRIBUTION   Virginia.
QUANTITY  2 units were distributed.
REASON    Blood products tested negative for the HIV-1 p24            antigen, but were collected
from a donor who previously
                         tested repeatedly reactive for the HIV-1
          p24 antigen,             neutralization indeterminate.

_______________
PRODUCT   Recovered Plasma.  Recall #B-1171-7. 
CODE Unit 3690607.
MANUFACTURER   Walter Reed Army Medical Center,             Washington, D.C.
RECALLED BY    Manufacturer, by telephone on or about July 20,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor with a history
          of cancer.

_______________
PRODUCT   Platelets.  Recall #B-1176-7.
CODE Unit 28108-6527.
MANUFACTURER   United Blood Services, San Angelo, Texas.
RECALLED BY    Blood Systems, Inc.,               Scottsdale,
               Arizona, by telephone on June 23, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Texas.
QUANTITY  1 unit was distributed.
REASON    Blood product corresponded to a unit of Red Blood Cells
          which was clotted.

_______________
PRODUCT   Red Blood Cells and Recovered Plasma.  
     Recall #B-1177/1178-7.
CODE Unit 20043-6257.
MANUFACTURER   United Blood Services, Lafayette, Louisiana.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated March 13, 1997.  Firm-initiated
               recall complete.
DISTRIBUTION   Louisiana and Switzerland.
QUANTITY  1 unit was distributed.
REASON    Blood products were collected from a donor who was
          previously deferred for an unacceptable medical
          history. 

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Recovered Plasma. 
          Recall #B-1181/1183-7.
CODE Unit 20030-6759.
MANUFACTURER   United Blood Services, Lafayette, Louisiana.

                               -9-RECALLED BY    Blood Systems, Inc.,               Scottsdale,
               Arizona, by letters dated June 19, 1997 and April
               9, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Louisiana, Nevada, North Carolina.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor who
          previously self-excluded. 

_______________
PRODUCT   Red Blood Cells and Recovered Plasma.
     Recall #B-1184/1185-7.
CODE Unit 49F18225.
MANUFACTURER   American Red Cross Blood Services, Tulsa,
               Oklahoma.
RECALLED BY    Manufacturer, by letter dated April 15, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Oklahoma and California.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor whose mother
          was diagnosed as a chronic hepatitis B carrier.

_______________
PRODUCT   Red Blood Cells and Recovered Plasma.
     Recall #B-1186/1187-7.
CODE Unit 49F94249.
MANUFACTURER   American Red Cross Blood Services, Tulsa,
               Oklahoma.
RECALLED BY    Manufacturer, by letters dated May 20 and 16,
               1997.  Firm-initiated recall complete.
DISTRIBUTION   Oklahoma and California.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor taking the
          drug      Proscar.

_______________
PRODUCT   Platelets.  Recall #B-1188-7.
CODE Unit 29442-9680.
MANUFACTURER   United Blood Services, Chicago, Illinois.
RECALLED BY    Manufacturer, by telephone on May 2, 1997, and by
               letter on May 22, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Illinois.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor who reported            travel to an area
designated as  endemic for malaria.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT   Platelet, Pheresis.  Recall #B-1064-7.
CODE Unit #40P51318.
                               -10-MANUFACTURER   American Red Cross Blood Services, Peoria,
               Illinois.
RECALLED BY    Manufacturer, by telephone on May 21, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Iowa.
QUANTITY  1 unit was distributed.
REASON    Unlicensed blood product was distributed in interstate
          commerce.

_______________
PRODUCT   Red Blood Cells.  Recall #B-1140-7
CODE Unit  49FJ56584.
MANUFACTURER   American Red Cross Blood Services, Tulsa,
               Oklahoma.
RECALLED BY    Manufacturer, by telephone on April 17, 1997., and
               by letter on May 6, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Oklahoma.
QUANTITY  1 unit was distributed.
REASON    Blood product, which was exposed to unacceptable
          shipping temperatures, was distributed. 

_______________
PRODUCT   Recovered Plasma.  Recall #B-1143-7.
CODE Units  49FJ57215 and 49FJ57223.
MANUFACTURER   American Red Cross Blood Services, Tulsa,
               Oklahoma.  
RECALLED BY    Manufacturer, by fax on June 16, 1997,  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY  2 units were distributed.
REASON    Blood products were collected from donors whose body
          temperature and arm inspection had not been documented. 

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Platelets,
          Pheresis; d) Platelets for further manufacture; e)
          Fresh Frozen Plasma; f) Cryoprecipitated AHF; g)
          Recovered Plasma.
     Recall #B-1151/1157-7.
CODE Unit numbers:  a) 16H98694, 16M53382, 16L66909, 16L42628,
     16M50423, 16H01422, 16K31854, 16K23045, 16H89762, 16H68912,
     16J43115, 16C20823, 16T43275, 16J31608, 16F90773, 16K30301,
     16L75232, 16F81105, 16H69681, 16J43222, 16G76169, 16H98667,
     16K05022, 16F90341, 16L69428, 16F90770, 16G13048; b)
     16H98694, 16M53382, 16L66909, 16M50423, 16K31854, 16J43115, 
     16T43275, 16F90773, 16L75232, 16J43222, 16G76169, 16H98667,
     16L69428, 16F90770; c) 16P20198;
                               -11-     d) 16F90341; e) 16H98694, 16M53382, 16C20823, 16H98667,
     16K05022; f) 16J31608, 16K30301, 16H69681; g) 16L66909,
     16L42628, 16M50423, 16H01422, 16K31854, 16K23045, 16H89762,
     16H68912, 16J43115, 16T43275, 16J31608, 16F90773, 16K30301,
     16L75232, 16F81105, 16H69681, 16J43222, 16G76169, 16F90341,
     16L69428, 16F90770, 16G13048. 
MANUFACTURER   American Red Cross Blood Services,                Columbus, Ohio.
RECALLED BY    Manufacturer, by letters dated December 6, 1995 or
               July 5, 1996, and on January 1, 1996.
DISTRIBUTION   Ohio, California, New York, Switzerland.
QUANTITY  a) 27 units; b) 14 units; c) 1 unit; d) 1 unit; e) 5
          units; f) 3 units; g) 22 units were distributed.
REASON    Blood products tested repeatedly reactive for the                hepatitis B surface
antigen (HBsAg), and were subjected
                         to additional retesting. 

_______________
PRODUCT   Fresh Frozen Plasma.  Recall #B-1169-7.
CODE Unit #1280144.
MANUFACTURER   Central Kentucky Blood Center, Lexington,
               Kentucky.
RECALLED BY    Manufacturer, by telephone on April 23, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Kentucky.
QUANTITY  1 unit was distributed.
REASON    Blood product, which tested initially reactive for the
          HIV-1   p24 antigen, and was not retested in duplicate. 

_______________
PRODUCT   Platelets.  Recall #B-1172-7.
CODE Unit numbers:  KZ36319, KZ36325-1, KZ36325-2, KZ35430,
     KZ36322-1, KZ36322-2, KZ35432. 
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by telephone on July 17, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Virginia.
QUANTITY  7 units were distributed.
REASON    Blood products were incorrectly tested for the HIV-1
          p24 antigen.

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Platelets,
          Pheresis.  Recall #B-1173/1175-7.
CODE Unit numbers:  a) 1210363, 1210376, 2681492, 2681527,
     2681484, 2681490, 2671498, 2681500, 2681501, 2681502,
     2681503, 2681504, 2681506, 2681507, 2681508, 2681511,
     2681512, 2681520, 2681532, 2681535, 2681536; b) 1210372;
                               -12-     c) 1210378, 1210381, 1210384, 1210389, 1210398.
MANUFACTURER   Houchin Community Blood Bank, Bakersfield,
               California.
RECALLED BY    Manufacturer, by telephone on July 29, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY  a) 21 units; b) 1 unit; c) 5 units were distributed.
REASON    Blood products were incorrectly tested for the HIV-1
          p24 antigen.


RECALLS AND FIELD CORRECTIONS -- DEVICES -- CLASS
II ========
_______________
PRODUCT   Radiation Protective Devices, under the label Wolf X
          Ray (Medical Division) or Flow X Ray (Dental Division:
     a) Pro-tek-to PAD, Stock numbers: 75055, 75057
     b) Apronettes, Stock numbers: 69095, 69096, 69097, 69098
     c) Gonad Shields, Stock numbers: 755000 (Beige)
     d) Thyroid Collars, 5x27
     e) Easy Wrap, Stock numbers:  85020, 65021, 65022, 65023,
     65024, 65025
     f) Special Procedure Apron, Stock numbers:          67088,
     67089
     g) OR Special Procedure Apron, Stock numbers: 75062, 75063
     h) Fr. Closing Spec. Proc. Apron, Stock numbers: 68090,
     68091, 68092, 
     i) Coat Type, Stock numbers: 63002, 63013, 63001, 63011,
     63074, 63075, 63015, 63016, 63071, 63073
     j) Conventional Type, Stock Numbers: 62007, 62017, 62008,
     62018, 62038, 62039, 62003, 62004
     k) Adult Bib Apron,  Stock number 75050
                  Child's Bib Apron, Stock number 75052
     l) Thyroid Collar Only, Stock Number 75056
     m) Child's Bib Apron w/collar, Stock Number 75053
     n) Adult Bib Apron w/Attached Thyroid Collar, Stock number
     75051
     o) Panoramic X-Ray Apron, Stock number 75061.
     Recall #Z-780/794-7.
CODE All aprons bearing the operator number dates between January
     23, 1997 through June 11, 1997.
MANUFACTURER   Wolf X-ray, West Hempstead, New York.
RECALLED BY    Manufacturer, by letter June 20, 1997.  Firm-initiated recall ongoing.
                               -13-DISTRIBUTION   Nationwide and international.
QUANTITY  FLOW X Ray labeled devices
     Bib                           555
     Child apron                   29
     Adult apron                   11
     Thyroid shield                37
     Panoramic apron               19
     Gonad Shield                  1
     Special Procedures Coat       2
     Easy Wrap apron               4
     Front Closing apron           1
     WOLF X Ray labeled devices
     Conventional apron            55
     Bib                           380
     Easy Wrap apron               341
     Coat apron                    168
     Special procedures apron      33
     Apronette                     131
     Wrap Around apron             29
     Pro Tek To Pad                20
     Gonad Shield                  112
     Child                         19.
REASON    The radiation protection devices contain lead
          contaminated with small amounts of radioactive
          substances.
______________
PRODUCT   Leaded Radiation Protective Devices:
        Product                     Model No. 
     a)   Mesh, Velcro Apron          MVA
        Mesh, Velcro Apron OR       MVA-OR
        Mesh Buckle Apron           MBA
        Velcro Apron                VA
        Buckle Apron                BA
        Tie Apron                   TA
        Half Apron                  HA
        Wrap-Around Apron           WA
     b) Skirt/Vest (2 Piece Set)    SV
     c) Thyroid Collar              TC
     d) Gonadal Shield, 3"x5"       GS35
        Gonadal Shield, 5"x7"       GS57
        Gonadal Shield, 7"x9"       GS79.
     Recall #Z-803/806-7.
CODE Recall covers the identified devices with  Quality Control
     Numbers beginning with "1405", "1443" or "1456". 
MANUFACTURER   The Kennedy Company, Scottsboro, Alabama (supplier
               leaded vinyl)
     Pro-Tech X-Ray, Inc., Tallassee, Alabama (finished product).
RECALLED BY    Pro-Tech X-Ray, Inc., Tallassee, Alabama, by
               telephone from July 3, 1997 through July 7, 1997. 
               Firm-initiated recall complete.
                               -14-DISTRIBUTION   Nationwide.
QUANTITY  200 devices were distributed.
REASON    The radiation protection devices contain lead
          contaminated with small amounts of radioactive
          substances.

_______________
PRODUCT   CRITIKON Complete Inflation Systems
          a) CRITIKON Complete Inflation System, 
     b) CRITIKON Soft Complete System 
     c) Complete Inflation System.
     Recall #Z-813/815-7. 
CODE Product Number     Lot Numbers
     a) 22810697        0997,
        22830067        0977, 0987
                  22850797        0987
     b) 25800537        0557, 0777, 0787,
                                  0797, 0807,
                  25810797        0807, 0847, 0917,
                                  0947
     c) 23800157        0437, 1077
                  23810167        0227, 0327, 0917,
                                  0927
                  23830147        0157, 0367, 0707,
                                  1017.
MANUFACTURER   Surgikos S.A. de C.V., Chihuahua, Mexico.
RECALLED BY    Johnson & Johnson Medical, Inc., Tampa, Florida,
               by letter on May 10, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide and Canada.
QUANTITY  7,185 devices were distributed; firm estimated that
          3,000 devices remained on market at a time of recall
          initiation.
REASON    The inflation systems have latex bulbs but are labeled
          as "latex free" on the devices.

_______________
PRODUCT   Stryker Instruments brand 1.5 mm Drill Bits:
     a) Short, Part #5110-100-15; b) Medium, Part #5120-100-15. 
     Recall #Z-818/819-7.
CODE Lot numbers:  a) 97051285; b) 97056265 and 97062255.
MANUFACTURER   Stryker Instruments, Division of Stryker
               Corporation, Kalamazoo, Michigan.
RECALLED BY    Manufacturer, by letter on July 2, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   California, Florida, Illinois, Louisiana,
               Maryland, Michigan, South Carolina, Texas,
               Wisconsin, Hong Kong.
QUANTITY  185 drill bits were distributed.
REASON    The drill bits may break during use at maximum speed,
          resulting in injury to the patient or user.
                               -15-_______________
PRODUCT   Couer PCA-10 Disposable Angiography Syringes a) Part
          No. C305-1035 (bulk, non-sterile);
     b) Part No. C305-0035 (individual, sterile).
     Recall #Z-820/821-7.
CODE All lot numbers.
MANUFACTURER   Coeur Laboratories, Inc., Raleigh, North Carolina.
RECALLED BY    Manufacturer, by letter on May 20, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY  a) 36,000 units; b) 5,400 units were distributed.
REASON    There may be air leakage into syringes during
          aspiration due to a poor seal between the barrel and
          the plunger.

_______________
PRODUCT   Flowselector, allows the user to switch between one of
          three ports for delivering oxygen without having to
          rearrange the tubing from the flowmeter:
     a) Precision Medical Label, Model No. PM1000;
     b) Medical Fittings Label, Model No. MF8025.
     Recall #Z-824/825-7.
CODE All products.
MANUFACTURER   Precision Medical, Inc., Northampton,
               Pennsylvania; Medical Fittings, Northampton,
               Pennsylvania.
RECALLED BY    Precision Medical, Northampton, Pennsylvania, by
               letter on or about July 29, 1997.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and Canada.
QUANTITY  Undetermined.
REASON    The complete directions for use are on the box the
          product is distributed in, but not on the product's
          immediate labeling.  The box and the directions are
          discarded after the device is installed.

_______________
PRODUCT   Radiation Protective Devices:
     a) Small Half-Apron, Model No. AHS-18, Regular Half-Apron,
     Model No. AHR-24;
     b) Thyroid Collar, Model No. LTC-60;
     c) Small Gonadal Shield, Model LGS-10, Medium Gonadal
     Shield, Model LGM-20, Large Gonadal Shield, Model No.
     LGL-30, Gonadal Shield Set  (1 of each size) Model No.
     LGA-50;
     d) Small/Medium Mitten, Model No. LMS-12, Large Mitten,
     Model No. LML-15;

                               -16-     e) Small Lead Blanket, Model No. LBC-18, Medium Lead
     Blanket, Model No. LBC-24, Large Lead Blanket, Model No.
     LBC-48.
     Recall Z-827/831-7.
CODE Serial numbers -- Model No. AHS-18:
     001365, 001366, 001367, 001399, 001404, 001405, 001408,
     001480, 001485, 001495, 001496, 001497, 001498, 001499,
     001500, 001506, 001507, 001508, 001509, 001510, 001511,
     001585, 001593, 001594, 001595, 001596, 001597, 001598,
     001599, 001600, 001601, 001602, 001628, 001629, 001630,
     001647.
     Model No. AHR-24:
     001368, 001369, 001370, 001381, 001382, 001406, 001407,
     001477, 001479, 001512, 001530, 001531, 001532, 001648,
     001649.
     Model No. LTC-60:
     001478, 001501, 001502, 001503, 001504, 001505, 001524,
     001548, 001549, 001550, 001551, 001552, 001553, 001554,
     001571, 001572, 001573, 001574, 001575, 001576, 001577,
     001578, 001579, 001580, 001581, 001582, 001586, 001591,
     001592, 001623, 001624, 001626, 001627, 001636, 001645,
     001646, 001655, 001656, 001657.
     Model No. LGS-10:
     001386, 001452, 001455, 001492, 001527, 001538, 001539,
     001540, 001541, 001542, 001543, 001555, 001558, 001562,
     001565, 001567, 001611, 001614, 001617, 001620, 001631,
     001638, 001642, 001666.
     Model No. LGM-20:
     001389, 001453, 001456, 001528, 001544, 001556, 001559,
     001563, 001568, 001583, 001589, 001590, 001603, 001604,
     001605, 001606, 001607, 001608, 001609, 001610, 001612,
     001615, 001618, 001621, 001625, 001632, 001637, 001639,
     001643, 001651, 001667.
     Model No. LGL-30:
     001376, 001377, 001380, 001387, 001388, 001392, 001393,
     001401, 001402, 001403, 001416, 001417, 001418, 001419,
     001420, 001424, 001425, 001426, 001427, 001428, 001429,
     001430, 001431, 001432, 001433, 001437, 001450, 001454,
     001457, 001458, 001459, 001460, 001487, 001488, 001493,
     001529, 001545, 001557, 001560, 001564, 001566, 001569,
     001587, 001588, 001613,      001616, 001619, 001622, 001633,
     001634, 001635, 001640, 001641, 001644, 001652, 001664,
     001665, 001668.
                               -17-     Model No. LGA-50:
     (Serial Nos. are those on the individual shield):  001383,
     001384, 001385, 001409, 001410, 001411, 001412, 001413,
     001414,      001438, 001439, 001440, 001441, 001442, 001443,
     001444, 001445, 001446, 001462, 001463, 001464, 001465,
     001466, 001467, 001468, 001469, 001470, 001471, 001472,
     001473, 001489, 001490, 001491, 001658, 001659, 001660,
     001661, 001662, 001663.
     Model No. LMS-12:
     001362, 001363, 001364, 001481, 001482, 001483, 001484,
     001561, 001650, 001653, 001654.
     Model No. LML-15:  001525, 001526, 001547.
     Model No. LBC-18:  001522, 001523, 001546.
     Model No. LBC-24:  001570.
     Model No. LBC-48:  001345, 001475, 001476, 001494.
MANUFACTURER   Techno-Aide, Division of National Service
               Industries, Inc., Nashville, Tennessee (Finished
               Devices). 
     The Kennedy Company, Scottsboro, Alabama (Supplier Leaded
     Vinyl).
RECALLED BY    Techno-Aide, Nashville, Tennessee, by letter dated
               July 14, 1997.  Firm-initiated recall ongoing.
QUANTITY  283 devices were distributed.
REASON    Radiation protection devices contain lead contaminated
          with small amounts of radioactive substances.

_______________
PRODUCT   Dimension Clinical Chemistry Systems, used for glucose
          testing:
     a) Model XL, All Serial Numbers manufactured  since 7/13/97;
     b) Model RxL, All Serial Numbers manufactured since
     10/21/96.  Recall #Z-832/833-7.
CODE a) All serial numbers manufactured since 7/13/95; b) All
     serial numbers manufactured since 10/21/96.
MANUFACTURER   Dade International, Newtown, Connecticut
               (instrument); Dade International, Newark, Delaware
               (reagent cartridge).
RECALLED BY    Dade International, Newark, Delaware, by letter
               dated May 7, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY  a) 231 units; b) 68 units were distributed.

                               -18-REASON    When the devices perform testing using Glucose Flex
          reagent cartridges, they report unflagged glucose
          results which can be as much as 40% below expected
          values.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT   AMRESCO LDH-L Reagent, Catalog No. 7711, for the
          quantitative determination of lactate dehydrogenase
          (LDH) activity in human serum.  Recall #Z-801-7.
CODE Lot No. 3017A67.
MANUFACTURER   AMRESCO, Inc., Solon, Ohio.
RECALLED BY    Manufacturer, by letter on July 18, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida and Texas.
QUANTITY  9 kits were distributed.
REASON    The device contains less of the powdered reagent than
          is labeled.

_______________
PRODUCT   1.3mm Ti Box Plate (Bone Fixation Plate), Part 421.095,
          an orthopaedic medical device used for bone fixation. 
          Recall #Z-808-7.
CODE Lot No. A3GL518.
MANUFACTURER   Synthes (USA), Monument, Colorado.
RECALLED BY    Synthes (USA), Paoli, Pennsylvania, by memorandum
               on November 12, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY  50 units were distributed.
REASON    The plates immediate container was identified with a
          green colored label instead of a yellow colored label.

_______________
PRODUCT   1.3mm Ti Oblique L-Plate (Bone Fixation Plate), Part
          421.052, an orthopaedic medical device used for bone
          fixation.
     Recall #Z-809-7.
CODE Lot No. A3GK392.
MANUFACTURER   Synthes (USA), Monument, Colorado.
RECALLED BY    Synthes (USA), Paoli, Pennsylvania, by memorandum
               on April 24, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Nationwide.
QUANTITY  Firm estimates none remains on the market.
REASON    The plates immediate container was identified with a
          green colored label, which is used for 1.0mm plates,
          instead of a gold colored label.

                               -19-_______________
PRODUCT   1.7mm Ti Emergency Screws with Fluted Tips, orthopaedic
          medical devices used for bone fixation:  a) 1.7mm Ti
          Emergency Screw (5mm), Part 400.655; b) 1.7mm Ti
          Emergency Screw (3mm), Part 400.653. Recall #Z-810/811-7.
CODE           Lot Numbers:  a) A3GJ390; b) A3GJ393. 
MANUFACTURER   Synthes (USA), Monument, Colorado.
RECALLED BY    Synthes (USA), Paoli, Pennsylvania, by memorandum
               on April 23, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Pennsylvania, Virginia, North Carolina, South
               Carolina, Kentucky, Florida, Texas, Minnesota,
               Wisconsin, Ohio, Kansas, California.
QUANTITY  Firm estimates none remains on the market.
REASON    The screws were anodized with a gold color instead of a
          teal color.

_______________
PRODUCT   1.5mm Ti L-Plate, Part 446.516, orthopaedic medical
          device used for bone fixation.
     Recall #Z-812-7.
CODE Lot No. A3IC949.
MANUFACTURER   Synthes (USA), Monument, Colorado.
RECALLED BY    Synthes (USA), Paoli, Pennsylvania, by memorandum
               on September 23, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Texas, Kansas, Missouri, Arkansas, Pennsylvania.
QUANTITY  Firm estimates none remains on the market.
REASON    The screws were anodized with a gold color instead of a
          teal color.

                               -20-

END OF ENFORCEMENT REPORT FOR AUGUST 13, 1997.  BLANK PAGES MAY
FOLLOW.

                                   ####

End of Enforcement Report for August 13, 1997
[FDA HOME PAGE]