FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

August 6, 1997                                        97-32

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT   Sea Legs brand Whole Leg Style Imitation King Crabmeat,
          in 5 pound clear plastic bags.
     Recall.  #F-568-7.
CODE 2 7134.
MANUFACTURER   Sugiyo USA, Inc., Anacortes, Washington.
RECALLED BY    Manufacturer, by letter on June 2, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Washington state.
QUANTITY  334 30-pound cases were distributed.
REASON    Product is contaminated with Listeria monocytogenes.

_______________
PRODUCT   Dips, salads, grape leaves and salsa under the Cedar's
          label:
     1.  Original Hommus Tahini 
     2.  Mediterranean Spice Hommus Tahini
     3.  No Salt Added Hommus Tahini
     4.  Baba Ghannouj
     5.  Roasted Baba Ghannouj
     6.  Garlic & Chive Hommus Tahini
     7.  Roasted Red Pepper Hommus Tahini
     8.  Roasted Almond Hommus Tahini
     9.  Pimento & Olive Hommus Tahini
     10. Sundried Tomato & Basil Hommus Tahini
     11. Sauteed Onions & Peppers Hommus Tahini
     12. Vegetable Hommus Tahini
     13. Garlic Lovers Hommus Tahini
     14. Beantown Sport Dip
     15. Garden Lentil Salad
     16. Taboule Salad 
     17. Grape Leaves Stuffed with White Rice 
     18. Boston Vegetable Salsa 
     Dips under the Mr. Hommus label:
     19. Original Hommus
     20. Salsa Hommus
     21. Garlic Hommus
     22. Dill Hommus
     23. Scallion Hommus
     24. Vegetable Hommus
     25. Roasted Garlic & Chive Hommus
     26. Roasted Red Pepper Hommus
     27. 40 (or Magic) Spice Hommus
     28. Sun-dried Tomato & Basil Hommus
     29. Sport Dip
     The following dips under the Vita Heavenly   Hommus! label:
     30. Original Hommus
     31. Dill Hommus
          32. Extra Garlic Hommus
     33. Vegetable Hommus
     34. Roasted Red Pepper Hommus
     35. Lemon Dill Hommus
     36. Spice Hommus
     37. HOT!!! Red Pepper Hommus
     38. Tomato & Basil Hommus
     39. Bagel with Chives Hommus
     40. Bagel Hommus
     The following dips and salads under the Sidon label
     (distributed in Canada only):
     41. Original Hommus Tahini
     42. Piment & Olive (Pimento & Olive)
     Hommus Tahini
     43. Pour Les Amants D'Lail (Garlic Lovers)
     Hommus Tahini
     44. Vegetable Hommus Tahini
     45. Baba Ghannouj
     46. Salad de Lentilles Jardiniere (Garden
     Lentil Salad)
     47. Salade Taboule.  Recall #F-569/615-7.
CODE All lots.
MANUFACTURER   Cedars Mediterranean Foods, Inc., Plaistow, New
               Hampshire.

                               -2-RECALLED BY    Manufacturer, by telephone beginning on June 12,
               1997, followed by fax.  Firm-initiated recall
               ongoing.  See Also FDA press release dated June
               12, 1997.
DISTRIBUTION   Hawaii, New Hampshire, Vermont, Connecticut,
               Massachusetts, Maryland, Maine, New York, New
               Jersey, Pennsylvania, Rhode Island, Illinois,
               North Carolina, South Carolina, Georgia, Virginia,
               Michigan, Minnesota, Ohio, Canada.
QUANTITY  1,166,857 units produced between April 1997 to June 12,
          1997 (including 8 ounce, 16 ounce, 5 pound).
REASON    The products may be contaminated with Listeria
          monocytogenes.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Fresh Frozen
          Plasma; d) Recovered Plasma.
     Recall B-1079/1082-7.
CODE Unit numbers:  KH71577, KH72926, KH74741; b) KH71577; c)
     KH71577; (d) KH72926, KH74741.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, telephone and fax on April 25, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Virginia, Illinois, California, Switzerland.
QUANTITY  a) 3 units; b) 1 unit; c) 1 unit; d) 2 units were
          distributed.
REASON    Blood products were collected from a donor with a
          history of cancer.

_______________
PRODUCT   a) Red Blood Cells; b) Recovered Plasma.
     Recall #B-1098/1099-7.
CODE Unit #36FK11300.
MANUFACTURER   American Red Cross Blood Services, Columbia, South
               Carolina.
RECALLED BY    Manufacturer, by letter dated April 11, 1997, and
               by fax on April 4, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   South Carolina and California.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor whose spouse
          has hepatitis C.

_______________
PRODUCT   Fresh Frozen Plasma, Pheresis.
     Recall #B-1100-7.
CODE Unit #4020498.
MANUFACTURER   Central Indiana Regional Blood Center, Inc.,
               Indianapolis, Indiana.

                               -3-RECALLED BY    Manufacturer, by telephone on January 17, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Indiana.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor with an
          elevated body temperature.

_______________
PRODUCT   Fresh Frozen Plasma.  Recall #B-1104-7.
CODE Unit #03GT25232.
MANUFACTURER   American Red Cross Blood Services, Atlanta,
               Georgia.
RECALLED BY    Manufacturer, by telephone on May 16, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Georgia.
QUANTITY  1 unit was distributed.
REASON    Blood product corresponded to a unit of Red Blood Cells
          that was reported as clotted.

_______________
PRODUCT   a) Platelets; b) Fresh Frozen Plasma.
     Recall #B-1105/1106-7.
CODE Unit #03GP11980.
MANUFACTURER   American Red Cross Blood Services, Atlanta,
               Georgia.
RECALLED BY    Manufacturer, by telephone on June 4, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Georgia.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were prepared from a unit of Whole Blood
          designated as a "difficult collection".

_______________
PRODUCT   Platelets.  Recall #B-1107-7.
CODE Unit #03FJ41109.
MANUFACTURER   American Red Cross Blood Services, Atlanta,
               Georgia.
RECALLED BY    Manufacturer, by telephone on June 3, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Georgia.
QUANTITY  1 unit was distributed.
REASON    Blood product was prepared from a unit of Whole Blood
          designated as a "difficult collection".

_______________
PRODUCT   Platelets.  Recall #B-1108-7.
CODE Unit #28GM03018.
MANUFACTURER   American Red Cross Blood Services, Savannah,
               Georgia.
RECALLED BY    Manufacturer, by telephone on or about March 3,
               1997.  Firm-initiated recall complete.
                               -4-DISTRIBUTION   South Carolina.
QUANTITY  1 unit was distributed.
REASON    Blood product was prepared from a unit of Whole Blood
          designated as a "difficult collection".

_______________
PRODUCT   Platelet Pheresis, Split unit.
     Recall #B-1109-7.
CODE Unit #21P67362.
MANUFACTURER   American Red Cross, Portland, Oregon.
RECALLED BY    Manufacturer, by letter dated May 20, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Oregon.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor who reported
          travel to an area designated as endemic for malaria.

_______________
PRODUCT   Platelet Pheresis, Irradiated.  
     Recall #B-1110-7.
CODE Unit #21P67582.
MANUFACTURER   American Red Cross, Portland, Oregon.
RECALLED BY    Manufacturer, by letter dated May 20, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Oregon.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor who reported
          travel to an area designated as endemic for malaria.

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Fresh Frozen
          Plasma; d) Recovered Plasma.
     Recall #B-1112/1115-7.
CODE Unit numbers:  a) 4002374, 7108543, 7407945, 7133679,
     7332989, 4008031, 1800905, 1820061, 7716673; b) 7108543,
     7407945, 4008031, 7716673; c) 7716673; d) 7108543, 7407945,
     7133679, 7332989, 4008031, 1800905, 1820061.
MANUFACTURER   LifeSource, Glenview, Illinois.
RECALLED BY    Manufacturer, by letters dated June 5 and 25,
               1997, July 9, 1997 or July 15, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Illinois, California, Arizona, District of
               Columbia.
QUANTITY  a) 9 units; b) 4 units; c) 1 unit; d) 7 units were
          distributed.
                               -5-REASON    Blood products tested negative for the antibody to the
          human immunodeficiency virus type 1 (anti-HIV-1), but
          were collected from a donor who previously tested
          repeatedly reactive for anti-HIV-1, Western blot
          negative.

_______________
PRODUCT   a) Red Blood Cells; b) Red Blood Cells, for further
          manufacture; c) Platelets; d) Platelets, for further
          manufacture; e) Fresh Frozen Plasma; f)
          Cryoprecipitated AHF; g) Recovered Plasma.  Recall #B-1116/1122-7.
CODE Contact FDA, Center for Biologics Evaluation and Research,
     Office of Compliance (301) 594-1191 for individual unit
     numbers recalled.
MANUFACTURER   American Red Cross Blood Services, Columbus, Ohio.
RECALLED BY    Manufacturer, by letters dated January 30, 1996,
               or July 5, 1996, and on February 1, 1996.  Firm-initiated recall complete.
DISTRIBUTION   Ohio, Michigan, Kentucky, Georgia, West Virginia,
               California, New York, Illinois, Switzerland.
QUANTITY  a) 53 units; b) 2 units; c) 20 units; d) 7 units; e) 9
          units; f) 10 units; g) 47 units were distributed.
REASON    Blood products tested repeatedly reactive for the
          antibody to the human immunodeficiency virus type 1
          (anti-HIV-1), and were subjected to additional
          retesting.

_______________
PRODUCT   a) Red Blood Cells; b) Red Blood Cells (Irradiated); c)
          Platelets; d) Platelets (Irradiated).  Recall #B-1130/1133-7.
CODE Unit numbers:  a) 1008883; b) 3000751; c) 1008883; d)
     3000751.
MANUFACTURER   Community Blood Center, Dayton, Ohio.
RECALLED BY    Manufacturer, by letter on February 18, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Ohio.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor who reported
          travel to an area designated as endemic for malaria.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT   Platelet Pheresis, Leukodepleted.  
     Recall #B-1066-7.
CODE Unit #40P53940.
                               -6-MANUFACTURER   American Red Cross Blood Services, Peoria,
               Illinois.
RECALLED BY    Manufacturer, by telephone on May 27, 1997, and by
               letter on June 9, 1997.  Firm-initiated recall
               complete.
DISTRIBUTION   Illinois.
QUANTITY  1 unit was distributed.
REASON    Unlicensed blood product was distributed without
          obliteration of U.S. license. 

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Recovered Plasma. 
          Recall #B-1070/1072-7.
CODE Unit numbers: a) 5047904; b) 5047904 and 5047905; c) 5047904
     and 5047905.
MANUFACTURER   Lorain County Blood Bank, Inc., Elyria, Ohio.
RECALLED BY    Manufacturer, by telephone on May 27, 1997,
               followed by letters dated May 28, 1997 or June 11,
               1997.  Firm-initiated recall complete.
DISTRIBUTION   Ohio and Kentucky.
QUANTITY  1) 1 unit; b) 2 units; c) 2 units were distributed.
REASON    Blood products were incorrectly tested for the
          hepatitis B surface antigen (HBsAg), and a subsequent
          donation which tested negative for HBsAg.

_______________
PRODUCT   Red Blood Cells.  Recall #B-1073-7.
CODE Unit numbers:  30167-6907, 30167-6925, 30167-6939, 30167-6943.
MANUFACTURER   United Blood Services, Bismark, North Dakota.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               telephone on September 20, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   South Dakota.
QUANTITY  4 units were distributed.
REASON    Unlicensed blood products were distributed labeled with
          the firm's license number.

_______________
PRODUCT   Red Blood Cells.  Recall #B-1076-7.
CODE Unit #9503178.
MANUFACTURER   LifeShare Blood Centers, Beaumont, Texas.
RECALLED BY    LifeShare Blood Centers, Shreveport, Louisiana, by
               telephone on February 13, 1997.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas.
QUANTITY  1 unit was distributed.
REASON    Blood product was distributed after receipt of post
          donation cold symptoms.

                               -7-_______________
PRODUCT   Red Blood Cells.  Recall #B-1078-7.
CODE Unit #30168-7277.
MANUFACTURER   United Blood Services, Bismarck, North Dakota.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               telephone on December 23, 1996.  Firm-initiated
               recall complete. 
DISTRIBUTION   North Dakota.
QUANTITY  1 unit was distributed.
REASON    Blood product remained at room temperature for an
          unacceptable length of time.

_______________
PRODUCT   Buffy Coat.  Recall #B-1083-7.
CODE Unit #KH74741.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by telephone on April 25, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Virginia.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor with a history
          of cancer.

_______________
PRODUCT   a) Platelets; b) Fresh Frozen Plasma.  
     Recall #B-1093/1094-7.
CODE Unit #KQ21662.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by telephone on or about December
               22, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Virginia.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor taking the
          antibiotic Cipro.

_______________
PRODUCT   Source Plasma.  Recall #B-1096-7.
CODE Unit #G-56799-081.
MANUFACTURER   Alpha Therapeutic Corporation, Charlotte, North
               Carolina.
RECALLED BY    Alpha Therapeutic Corporation, Los Angeles,
               California, by letter on September 26, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Spain.
QUANTITY  1 unit was distributed.
REASON    Blood product tested positive for syphilis by RPR.

_______________
PRODUCT   Red Blood Cells.  Recall #B-1097-7.
CODE Unit #18Q91042.

                               -8-MANUFACTURER   American Red Cross Blood Services, Lansing,
               Michigan.
RECALLED BY    Manufacturer, by letter dated March 7, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Michigan.
QUANTITY  1 unit was distributed.
REASON    Blood product was collected from a donor whose
          screening was inadequately performed.

_______________
PRODUCT   a) Red Blood Cells; b) Platelets; c) Fresh Frozen
          Plasma.  Recall #B-1101/1103-7.
CODE Unit #KE53396.
MANUFACTURER   Virginia Blood Services, Richmond, Virginia.
RECALLED BY    Manufacturer, by letter dated December 9, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Virginia and Utah.
QUANTITY  1 unit of each component was distributed.
REASON    Blood products were collected from a donor taking the
          drug, Mellaril.

_______________
PRODUCT   Red Blood Cells. Recall #B-1111-7.
CODE Unit #21GJ79343.
MANUFACTURER   American Red Cross Blood Services, Portland,
               Oregon.
RECALLED BY    Manufacturer, by telephone on May 8, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Oregon.
QUANTITY  1 unit was distributed.
REASON    Blood product was labeled with an extended expiration
          date.

_______________
PRODUCT   Red Blood Cells.  Recall #B-1128-7.
CODE Unit numbers:  16LN06282, 16LR08707, 16LY18116.
MANUFACTURER   American National Red Cross, Columbus, Ohio.
RECALLED BY    Manufacturer, by letter on December 10, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Ohio and American Samoa.
QUANTITY  3 units were distributed.
REASON    Unit of pooled platelets implicated in a post
          transfusion reaction were found to be contaminated with
          Proprionibacterium acnes.

_______________
PRODUCT   Red Blood Cells Leukodepleted.  
     Recall #B-1129-7.
CODE Unit #16LS10360.
MANUFACTURER   American National Red Cross, Columbus, Ohio.
RECALLED BY    Manufacturer, by letter on October 10, 1996. 
               Firm-initiated recall complete.
                               -9-DISTRIBUTION   Ohio.
QUANTITY  1 unit was distributed.
REASON    Red Blood Cells were leukodepleted by filtration
          greater than five days past the time of collection.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS
II =========
_______________
PRODUCT   Reusable POLYCARBONATE DOMES without Diaphragm, Catalog
          No. TA1010"D", used to help measure physiological
          pressure by serving as a reservoir for sterile liquid
          and converts the incoming pressure into an electrical
          output for the monitor.  Recall #Z-690-7.
CODE      Lot Nos. SD1206, SD1207, SD1223, SD1286, SD1302,
          SD1342.
MANUFACTURER   Ohmeda (Singapore), Singapore.
RECALLED BY    Ohmeda Medical Devices Division, Inc., Murray
               Hill, New Jersey, by letter dated January 28,
               1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Florida, Alabama, Washington state, New Jersey,
               Illinois, California.
QUANTITY  2,238 pieces were distributed.
REASON    The lots incorrectly contain the suffix "D" which
          indicates a single use, sterile device, when in fact
          they are reusable non-sterile devices.

_______________
PRODUCT   ECOSYSTEM Endoscopic Ligation Cartridges, used as a
          part of the Weck Hemoclip Ecosystem Ligation System: 
          a)  Catalog No. 140145;
          b) Catalog No. 140245; c) Catalog No. 140601;
     d) Catalog No. 140603; e) Catalog No. 140604.
     Recall #Z-692/696-7.
CODE All Lots.
MANUFACTURER   Pilling Weck, Inc., Research Triangle Park, North
               Carolina.
RECALLED BY    Manufacturer, by letter on May 27, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY  48,914 products were distributed.
REASON    Products are subject to improper clip closure.

_______________
PRODUCT   InstaTrak Systems containing an Electromagnetic
          Tracking Device, used for the acute and chronic
          sinusitis endoscopic, dacryocystorhinostomy, optic
          nerve and orbital decompression, the removal of polyps,
          the

                               -10-     biopsy and removal of tumors, and the repair of CSF leaks,
     pituitary disorder, and encephalocele:
     a) InstaTrak System, Catalog No. IT-100ENT, an Image Guided
     Surgery System that employs a computer, software and
     electromagnetic tracking device that is used for endoscopic
     sinus surgery; b) ConneCTstat Workstation, Catalog No.
     CS-100, an independent workstation used to archive CT images
     onto a cartridge which maybe used on the InstaTrak System
     during surgery.  It does not have a position sensing and
     tracking capabilities.   
     Recall #Z-751/752-7.
CODE      Serial Nos. a) IT-10012, IT10013, IT10015,   IT10016,
          IT10020, IT10021, IT10022, IT10024,   IT10027, IT10029;
          b) CS101, CS102.
MANUFACTURER   Visualization Technology, Inc., Woburn,
               Massachusetts.
RECALLED BY    Manufacturer, by letter June 13, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   New York, Tennessee, Minnesota, Ohio, District of
               Columbia, Massachusetts, Illinois, Pennsylvania.
QUANTITY  (a) 10 units; b) 2 units were distributed.
REASON    The firm was found to be not in conformance with the
          Good Manufacturing Practice (GMP) for Medical Devices
          Regulation, in that it failed to provide adequate
          documentation the Electromagnetic Tracking Device (P/N
          1000045-001) met the firm's specifications.

_______________
PRODUCT   QuickStripe Chlamydia Ag, Catalog No. 1185/0185, a
          rapid, visual assay for the detection of Chlamydia
          trachomatis antigen in endocervical, urethral, or male
          urine specimens.    Recall  #Z-753-7.
CODE Kit Lot No. 0005-1.
MANUFACTURER   Chembio Diagnostics Systems, Inc., Medford, New
               York.
RECALLED BY    Diatech Diagnostics, Inc., Boston, Massachusetts
               (distributor), by letters on May 8, 1997, and June
               3, 1997.  Firm-initiated recall ongoing.
DISTRIBUTION   Russia, Egypt, Miami (Trinidad).
QUANTITY  302 kits (20 tests per kit) were distributed.
REASON    The cassette has a shortened "lip" for holding the test
          membrane in place, making it difficult to read the
          results.

                               -11-_______________
PRODUCT   Revolution Manual Wheelchair.  
     Recall #Z-768-7.
CODE Various serial numbers.
MANUFACTURER   Quickie Designs, Inc., Fresno, California.
RECALLED BY    Manufacturer, by letter on June 12, 1997.  Firm-initiated recall complete
DISTRIBUTION   Nationwide and international.
QUANTITY  2,749 units were distributed.
REASON    The Caster Hinge and Bolt may potentially detach
          causing the wheelchair to collapse and the user to fall
          and be injured.

_______________
PRODUCT   Lead Aprons, Vests, Wraps, Thyroid Collars and Gonad
          Shields, radiation protective devices:
     a) Aprons; b) Apronettes; c) Mini Aprons; d) Apron with
     Thyroid Collar; e) Wraps; f) Jackets; g) Vests; h) Skirt; i)
     Vest and Skirt; j) Thyroid Collars; k) Gonadal Shields. 
     Recall #Z-769/779-7. 
CODE a) Aprons
     E-Z-EM, Inc. Catalog Numbers -- Lot Numbers:
     A241 - Lot 132; A25F - Lot:  274; 
     A25H - Lot: 130, 131; A25K - Lots: 123, 131, 132, 134; A32J
     - Lots: 104, 186; A32K - Lot:  104; A32M - Lot: 186; A33F -
     Lot: 113; A33H -   Lot: 113; A33K - Lot: 93677; A33M -
     Lots:113, 284; A33S - Lot: 285; A35F - Lots: 123, 134, 274;
     A35G - Lots: 113, 134, 276; A35J - Lots:  113, 123; A35K -
     Lots: 102, 113,130, 132, 133, 134, 144;  A35L - Lots: 113,
     130;  A35M - Lots: 113, 130, 284; A35S - Lots: 113, 130 285;
     A43H - Lot: 286; A44H - Lot: 112; A44K - Lot: 186; A45K -
     Lots:  285, 286; A47G - Lots  102, 113, 123, 131, 165; A47H
     - Lots: 102, 113, 116, 126, 285, 132625; A47K - Lots: 111,
     113, 123, 126, 130, 131; A77G - Lot: 134; AF32G - Lot: 104;
     AF35G - Lots:  102, 104, 113, 116, 123, 126, 236; AF35H -
     Lots: 104, 113, 116, 131, 230; AF35J - Lots: 108, 111, 116,
     123, 133; AF35K - Lots: 113, 116, 123, 130, 131, 132, 133,
     134; AF35L - Lots:  102, 108, 116, 133, 276; AF35M - Lots:
     111, 116, 133, 134, 284; AF35S - Lots: 116, 130, 133, AF51K
     - Lots: 198, 285; AF52K - Lots: 274, 277; AF52S - Lot: 277;
     AF54H - Lot: 134; 
     AF54K - Lots: 130, 131; AF54M - Lots: 133, 134, 269; AF54S -
     Lot: 269;  AF55K - 130, 277, AF55S - Lot: 129, AF73S - Lot:
     285, 
     AF75G - Lot: 104, AF75K - Lot: 104, 
     AF77G - Lots:  102, 123, 129, 165, 285, 

                               -12-     AF77H - Lots:  116, 123, 126, 165, 191, 
     AF77J - Lot: 285, AF77L - Lot: 144, 
     AF83J - Lot: 123, AF85F, Lot:  113, 
     AF85G - Lot: 116, AF85H - Lot: 131; AF85J - Lots:  123, 133;
     AF85K - Lots: 113, 123, 130, AF85L - Lot: 113; AF85M - Lot:
     116; AF85S - Lot: 134; AF85T - Lot: 113; AF95F - Lot: 111,
     AF95G - Lots:  111, 129, 276; AF95H - Lots: 102, 116, 130;
     AF95K - Lots:  102, 111, 113, 123, 130, 134, 144, 234, 236;
     AF95L - Lots: 111, 116, 126, 130, 133, 276; AF95M - Lots: 
     112, 113, 130 133; AF95S - 116, 130, 133, 134, 135, 
     A900 - Lots:  113, 116, 133, 134, 144.
     Picker Sentry Lite, Catalog Numbers, Lot Numbers:
     253176 - Lots:  113, 116,133, 134, 274;
     253177 - Lots:  102, 111, 113, 116, 129, 130, 131, 134, 
     274; 253178 - Lots: 108, 116, 130, 274; 253179 - Lot: 274;
     253180 - Lots:  113, 116, 129, 131, 134; 253181 - Lots: 
     113, 123, 131, 144; 253369 - Lots:  116, 131, 132, 134, 138,
     165, 276, 288; 253370 - lots:  113, 116, 123, 134, 276, 288;
     253371 - Lots:  113, 116, 123, 129, 131, 276, 288; 253376 -
     Lots:  133, 134; 253377 - Lots: 130, 133, 134; 253378 -
     Lots: 131; 253379 - Lots: 131, 134; 256008 -Lots: 130, 136;
     256009 - Lots: 131, 177; 256223 - Lots: 131, 133; 256227 -
     Lots:  116, 131, 134; 256228 - Lots:  113, 116, 276; 256229
     - Lots:  133, 276; 256233 - Lot: 961091; 256236 - Lot: 116;
     256237 - Lots: 116, 129, 130, 131, 133, 134; 256238 - Lots:
     111, 113, 131, 132, 134, 165; 256239 - Lots: 131, 132, 133;
     256253 - Lots: 134, 274, 276; 256254 - Lots: 134, 234;
     256255 - lots: 113, 116, 130, 134, 276; 256258 - Lots: 111,
     116, 134; 256259 - Lots: 111, 133, 234, 288; 256260 - Lots: 
     111, 113, 116, 129, 130, 133, 134, 276,  286; 256261 - Lots:
     113, 133, 276, 280, 286;
     257882 - Lots: 108, 129, 133, 276; 258042 - Lot:  286;
     258043 - Lots: 104, 116, 165, 280;
     258078 - Lots: 130, 134, 286; 258080 - Lots:  113, 129, 130,
     133, 288; 258200 - Lots:  116, 126; 258889 - Lot: 116;
     258891 - Lots:  104, 108; 258892 - Lot: 133; 258894 - Lots: 
     104, 131; 258895 - Lots: 104, 108, 113, 116, 123, 130, 134;
     258896 - Lots: 104, 108, 116, 126, 130; 258897 - Lots: 108,
     113, 134, 276; 259249 - Lot: 144; 261871 - Lot: 286; 261872
     - Lot:  274; 261876 - Lot: 144; 261885 - Lot: 134;
     263694 - Lots: 126, 129, 131, 133;

                               -13-     263695 - Lots: 113, 129, 132, 133;
     264622 - Lots: 134, 284; 264624 - Lot: 105;
     264625 - Lots: 111, 116, 131, 133, 223, 234, 284, 286;
     264626 - Lots: 102, 130, 276;
     264627 - Lots: 131, 165, 236;
     264628 - Lots: 111, 131, 234;
     264629 - Lots: 113, 144
     264630 - Lots: 129, 130, 134, 165
     264631 - Lot: 276; 264632 - Lot: 131
     264633 - Lots: 113, 144
     264634 - Lots: 105, 113, 116, 129, 276
     264635 - Lots: 130, 276
     264636 - Lots: 113, 116, 129, 131, 133, 276
     264637 - Lots: 129, 133, 134, 274, 276, 286
     264638 - Lots: 113, 116, 129, 276
     264639 - Lots: 131, 136, 276, 286
     264709 - Lot: 136; 265287 - Lots: 113, 133
     265288 - Lots: 129, 276; 265289 - Lots: 133, 276; 265492 -
     Lots:  116, 131, 276
     265493 - Lot: 129; 265494 - Lot: 134
     265495 - Lots: 136, 138, 190
     265496 - Lots: 136, 177l; 265497 - Lots: 136,
     138; 265513 - Lots:  104, 105, 108, 116, 126, 130, 131, 134;
     136, 190, 223, 234, 276
     265579 - Lots: 140, 190; 266191 - Lot: 116
     266192 - Lot: 134; 266193 - Lot: 134
     266194 - Lot: 131; 266195 - Lot: 104
     266197 - Lot: 116, 131, 165
     266198 - Lot: 136; 266228 - Lot: 116, 123, 236
     266229 - Lot: 113, 134, 62694
     266230 - Lot: 131; 266231 - Lots: 104, 126, 137; 266232 -
     Lots: 108, 116
     266233 - Lots:  108, 144
     266236 - Lots: 144, 165, 236
     266238 - Lot: 104; 266239 - Lot: 234
     266240 - Lot: 126; 266241 - Lot: 294
     266245 - Lots: 144, 223, 234
     266246 - Lot: 126; 266305 - Lot: 105
     269682 - Lot: 104; 269683 - Lots: 104, 111, 113, 129, 130,
     133, 234, 276
     269684 - Lots:  111, 113, 116, 130, 131, 133, 134, 235, 284,
     286, 294
     269685 - Lots:  104, 111, 113, 129, 130, 234
     269686 - Lots:  111, 131, 234, 276
     269687 - Lots:  104, 111, 113,116, 130, 131, 234, 276, 284
     269688 - Lots:  116, 130, 133, 134, 234, 276
     269689 - Lots:  104, 111, 113, 116, 130, 131, 134, 223, 234,
     223, 276, 284, 286
     269690 - Lots:  136, 190
     400802 - Lots:  102
     400803 - Lots:  102, 113, 234

                               -14-     400804 - Lots:  130, 234
     400805 - Lots:  286; 264625T - Lot: 144
     264626T - Lot:  144; 264628T - Lots: 144, 223
     264634T - Lots: 105, 126
     264635T - Lots: 105, 126; 264636T - Lot: 126
     264637T - Lot:  126, 144; 
     264638T - Lots: 104, 236; 264639T - Lot: 102
     265287T - Lots: 108, 126; 265289T - Lot: 126
     265493T - Lot:  126; 265497T - Lots: 190, 191
     266195T - Lot:  104; 266197T, Lot:  104, 126
     269683T - Lots: 105, 126, 144
     269684T - Lots: 102, 108, 126, 144, 236
     269685T - Lots: 108, 144, 236, 260
     269686T - Lots: 104, 126, 144, 223
     269687T - Lots: 126, 144, 223, 236
     269688T - Lots: 104, 260
     269689T - Lots: 104,108, 126, 223, 260, 276
     269690T - Lot:  190; 9EZM79J - Lot: 116
     9EZM79K - Lots: 113, 130, 131, 132, 133, 280
     9EZM79L - Lots: 116, 134
     9EZM79S - Lots: 116, 284; EC94 - Lot: 133
     b) Apronettes
     E-Z-EM Catalog Numbers - Lot Numbers:
     A15A - Lots: 113, 133, 134
     A15B - Lots:  113, 133, 134, 286
     A15C - Lots:  123, 130, 134
     A15D - Lots:  113, 116, 130, 134, 286
     A15E - Lots:  113, 116, 123, 130, 132, 133, 230, 274
     c) Mini Aprons
     Picker Catalog Numbers - Lot Numbers:
     258162 - Lots: 113, 116, 129, 131, 132, 133, 134, 288
     258163 - Lots:  113, 116,129, 130, 131, 132, 133, 134, 223,
     280, 286, 288
     258164 - Lots:  113, 116, 123, 129, 130, 131, 132, 133, 223,
     277, 286, 288
     258165 - Lots:  113, 116, 123, 129, 132, 133, 134, 288;
     259356 - Lots: 113, 130
     d) Apron with Thyroid Collar
     Picker Catalog Numbers - Lot Numbers:
     253372 -  Lot: 130; 253373, Lot: 134
     256017 - Lots:  116, 276
     256018 - Lots:  113, 116, 130, 133, 286
     256019 - Lots:  116, 130, 134, 280, 286
     256020 - Lots:  116, 133
     256021 - Lots:  116, 131, 133
     256022 - Lots:  131,133
     265514 - Lots:  165, 177
     265515 - Lots:  104, 126, 236
     266247 - Lots:  104,126,144, 165, 236
     266248 - Lot:   236; 266249 - Lot: 105

                               -15-     e) Wraps
     E-Z-EM Catalog Numbers - Lot Numbers:
     AFS7L - Lots:  134, 274, 091594
     AFS7M - Lots:  126, 134
     AS7L - Lot: 134; AS7M - Lots: 116, 134
     Picker Catalog Numbers - Lot Numbers:
     256241 - Lot: 129;  256242 - Lot: 131
     256243 - Lot: 134; 256246 - Lot: 130
     265501 - Lot: 104, 113,127, 130, 134, 165, 223
     265503 - Lot: 92877
     265504 - Lots:  104, 105, 126, 127, 129, 134, 144, 165, 223,
     233, 236, 274
     265505 - Lots:  105, 108, 111, 116, 127, 134
     265506 - Lots:  129, 144
     265510 - Lots:  102, 104, 116, 123, 126, 127, 134, 136, 144,
     165, 223, 236
     265511 - Lots:  104, 123, 165
     265512 - Lots:  144, 236
     265517 - Lots:  126, 131
     265727 - Lots:  123, 134, 223
     265729 - Lots:  116, 123
     266203 - Lot:   104
     266204 - Lot:   129
     266205 - Lots:  108, 116, 234
     266206 - Lot:   116
     f) Jackets
     Picker Catalog Numbers - Lot Numbers:
     264640 - Lots:  104, 105, 108, 113, 116, 130, 134, 165, 223,
     236, 276
     264641 - Lot:  111; 264642 - Lot:  289
     264643 - Lots:  104, 105, 126, 134, 165, 223, 233, 236
     264644 - Lots:  104, 105, 108, 116, 126, 134, 223; 264645 -
     Lot: 144
     264649 - Lots:  104, 111, 116, 126, 127, 133, 136, 144, 223,
     236
     264650 - Lots:  104,123, 165
     264651 - Lots:  123, 144, 236
     265498 - Lots:  123, 129, 134, 144,223, 234
     265516 - Lots:  108, 126, 144, 165, 236
     266199 - Lots:  104, 081095
     266200 - Lots:  129, 234
     266201 - Lots: 108, 116
     266202 - Lot: 165
     g) Vests
     E-Z-EM Catalog Numbers - Lot Numbers:
     AFV7L - Lot:  134
     AFV7M - Lots:  126, 134
     AFV7S -  Lot: 123; AV7L, Lot:  134
     AV7M - Lots:  116, 134
     Picker Catalog Numbers - Lot Numbers:
     266526 - Lot: 104; 266527 - Lot:  104

                               -16-     266528 - Lots: 108, 130
     266529 - Lots: 108, 111
     266530 - Lot: 108; 266532 - Lot: 102
     266542 - Lot 108; 266543 - Lot 108
     266550 - Lot 130
     253369V - Lots: 113, 129, 131, 276
     253370V - Lots:  113, 123, 129, 131, 134
     253371V - Lots: 129, 131, 133, 134, 276
     h) Skirt
     Picker Catalog Numbers - Lot Numbers:
     266535 - Lots: 104, 130
     266536 - Lots:  102, 108
     266537 - Lot: 108; 266538, Lot: 108
     266540 - Lot: 111; 266547, Lot: 108
     266549 - Lot: 102; 266551 - Lot: 144
     I) Vest and Skirt
     Picker Catalog Numbers - Lot Numbers:
     256247 - Lots:  104, 116, 133, 134
     256248 - Lots:  102, 104, 116, 129, 133, 134
     256249 - Lots:  102, 108
     256251 - Lot:  116
     256252 - Lots:  130, 131, 133
     257896 - Lots:  104, 108, 144
     261953 - Lots:  135, 144
     261954 - Lots:  123
     j) Thyroid Collars
     E-Z-M Catalog Numbers - Lot Numbers:
     A656; 5x25 - Lots:  102, 113, 116, 123, 129, 130, 131, 132,
     133, 134, 144, 165, 236, 274,   276, 277, 284, 285, 286,
     294, 89993
     Picker Catalog Numbers - Lot Numbers:
     254179 - Lots:  104, 113, 116,, 123, 129, 130, 131, 132,
     133, 134, 140, 246, 274, 276, 284, 286, 288
     254180 - Lots:  113, 116, 129, 130, 133, 134, 276, 286
     258419 - Lots:  108, 113, 116, 123, 129, 130, 131, 133, 134,
     223, 276, 280, 284, 286
     264736 - Lot:   280
     265544 - Lots:  113, 116, 123, 126, 129, 130, 131, 133, 236,
     276, 280, 284
     265545 - Lots:  113, 116, 123, 129, 130, 131, 133, 144, 234,
     276
     265546 - Lots: 113, 116, 125, 129, 131, 239, 269
     266207, Lots:  116, 123, 130, 131, 134, 144, 165
     k) Gonadal Shields
     Picker Catalog Numbers, Lot Numbers:
     254178 - Lots:  110, 113, 116, 117, 118, 119, 128, 129, 131,
     133, 136, 144, 145, 168, 169, 180, 181, 191, 197, 198, 199,
     202, 206, 207, 

                               -17-     210, 213, 216, 224, 230, 236, 237, 246, 269, 277, 285, 286,
     1144, 8293, 8693, 8994, 9213, 9393, 10496, 62094, 62294,
     91594, 92093, 92194, 92328, 92329, 92334, 92339, 92387,
     92430, 92436, 92437, 92438, 92549, 92663, 92772, 92872,
     93221, 933326, 93328, 93343, 93443, 93449, 93450, 93671,
     93673, 95449, 101493, 101493,  121693, 921197, 1074205,
     1075215, 1076233, 1077258, 1079266, 1081300, 1132967,
     9112344, 9211105, 9211106, 9212107, 9212108, 9212109,
     11381011, 11381014, 11381020, 11431069, 011094, 92-10-94,
     92-11-97, 98-11-97
     262004 - Lots:  113, 116, 117, 119, 128, 129, 131, 132, 133,
     135, 141, 144, 162, 168, 179, 180, 183, 191, 197, 198, 202,
     207, 210, 213, 224, 236, 248, 269, 285, 286, 8994, 9213,
     11294, 92194, 92436, 92549, 92664, 92668, 92772, 92872,
     92876, 93443, 93892, 1075213, 1076213, 1132971, 1137999,
     11381020, 011094.
MANUFACTURER   E-Z-EM, Inc., Westbury, New York.
RECALLED BY    Manufacturer, by letters on June 6, 9, 20, 1997. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY  E-Z-EM & Picker Int'l labeled Aprons, Vests, and Wraps:
          a total of 2270 units (domestic & foreign combined) 
     E-Z-EM labeled Thyroid Collars: 560 units
     Picker International Thyroid collars:        1014 units (930
     domestically, 84 foreign)
     Picker International Gonad Shields: 694 units (681
     domestically, 13 foreign).
REASON    The radiation protection devices contain lead
          contaminated with small amounts of radioactive
          substances.

_______________
PRODUCT   Pilling Weck Manual Endoscopic Hemoclip Applier,
          Catalog #113511, for use in endoscopic application of
          Pilling Weck Hemoclips.  Recall #Z-795-7.
CODE All lot numbers.
MANUFACTURER   Hauser Medizintechnik, Gosheim, Germany (units
               date stamped 1/96 - 1/97); Edlo, SA Productos
               Medicos, Canoas, Brazil (units marked "Brazil").
RECALLED BY    Pilling Weck, Inc., (now Weck Closure Systems),
               Research Triangle Park, North Carolina, by letters
               on December 20, 1996, and May 7, 1997.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide, Mexico, Brazil, Peru, Puerto Rico,
               Canada, Argentina.

                               -18-QUANTITY  1,034 products were distributed.
REASON    The luer lock can become loose and/or the rotation knob
          may experience cracking which could affect the intended
          function of the applier.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT   STA-PTT-LT(5) Catalog No. 00668.  The kit provides
          reagent for the determination of the activated partial
          thromboplastin time (APPT) which is used as a general
          coagulation screening test.  Recall #Z-743-7.
CODE      Lot Nos. 951178, 951300, 962051, 962681, 963101, and
          963382.
MANUFACTURER   Diagnostica Stago, Franconville, France.
RECALLED BY    American Bioproducts Company, Parsippany, New
               Jersey,, by letter sent on February 24, 1997. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY  LOT #        AMT SHIPPED (BOXES) 
     951178         740       
     951300         235       
     962051         293       
     962681         96       
     963101         23       
     963382         51.   
REASON    Product is contaminated by Penicillium. 


RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS I ===========
  _______________
  PRODUCT Purina Sweetina Horse Feed (nonmedicated), in 50
            pound bags.  Recall #V-037-7.
  CODE    Lot #3525.
  MANUFACTURER Purina Mills, Inc., Macon, Georgia.
  RECALLED BY  Manufacturer, by telephone on February 10,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Georgia and Alabama.
  QUANTITY     113 bags were distributed.
  REASON  Feed was likely contaminated with medicated feedlot
            feed containing monensin and tylosin. 
  
  
  MEDICAL DEVICE SAFETY ALERTS:  ==============================
  _______________
  PRODUCT Datascope-Cardiac Assist Insertion Kits for use
            with Datascope Intra-Aortic Balloons and Datascope-Cardiac Assist Introducer Single Pack Kits for use
            with Datascope Inra-Aortic Balloons.  Safety
            Alert #N-016/027-7.
  
                               -19-CODE 1.  Model No. 0684-00-0187-01, Lot No. UN;
     2.  Model No. 0684-00-0257, Lot Nos. TZ, UO;
     3.  Model No. 0684-00-0264, Lot Nos. UF, UC, TW, VE, VF,
       VW, VZ;
     4.  Model No. 0784-00-0114, Lot Number None indicated;
     5.  Model No. 0784-00-0125, Lot Nos. VG, VP;
     6.  Model No. 0684-00-0251-11, Lot Nos. UE, UD, UJ,
       UL,VS, VY;
     7.  Model No. 0784-00-0103-04, Lot No. UA;
     8.  Model No. 0784-00-0128, Lot Nos. VA, VN;
     9.  Model No. 0684-00-251-13, Lot Nos. QU, UH, UG, UB,
       TY, TX, TU, UI, UP, VC, VD, VI;
     10.  Model No. 0784-00-0129, Lot Nos. VB, VQ;
     11.  Model No. 0784-00-0103-13, Lot No. UM;
     12. Model No. 0684-00-0208, Lot No. UQ.
  MANUFACTURER Datascope, Cardiac Assist Division, Fairfield,
                 New Jersey.
  Alerted By   Manufacturer, by letters on September 27,
                 1996, October 2 and 3, 1996.
  DISTRIBUTION Nationwide and international.
  QUANTITY     27,856 units were distributed.
  REASON  The inadvertent removal of the hemostasis valve
            which may lead to the unanticipated blood loss to
            the patient.  Notification also included
            instructions regarding use of subject device, and
            how to minimize bleeding if hemostasis valve is
            removed.
  
  END OF ENFORCEMENT REPORT FOR AUGUST 6, 1997.  BLANK PAGES MAY FOLLOW.