FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.
November 10, 1999                                             99-45

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS I

_______ PRODUCT Vanilla Swiss Almond Rice Dream Non-Dairy Dessert, packaged in pint containers. Recall #F-013-0. CODE 6-11877541 9147 BH. MANUFACTURER Humboldt Creamery Association, Fernbridge, California. RECALLED BY Imagine Foods, Inc., Palo Alto, California, by letter faxed on September 22, 1999, and by press release on September 23, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and Canada. QUANTITY 6,050 cases (12 packages per case) were distributed. REASON Product may be contaminated with peanut residue. ________ PRODUCT Ice Cream as follows: a) Chocolate Chip Cookie Dough Ice Cream in 1/2 gallon and 2 1/2 gallon size containers b) Pistachio Nut Ice Cream in 1/2 gallon and 2 1/2 gallon containers c) Rocky Road Ice Cream in 1/2 gallon and 2 1/2 gallon containers d) Tin Roof Ice Cream in 2 1/2 gallon size containers e) Caramel Cashew Ice Cream in 2 1/2 gallon size containers f) Turtle Sundae Ice Cream in 1/2 gallon and 2 1/2 gallon size containers g) Butter Pecan Ice Cream in 2 1/2 gallon size containers h) Chocolate Butter Pecan Ice Cream in 2 1/2 gallon size containers i) Maple Nut Ice Cream in 1/2 gallon and 2 1/2 gallon size containers j) Peanut Butter Cup Ice Cream in 2 1/2 gallon size containers k) Peanut Butter Ripple Ice Cream in 2 1/2 gallon size containers l) Chocolate Peanut Butter Ripple Ice Cream in 2 1/2 gallon size containers. Recall #F-015/026-0. CODE Products are not coded. All products distributed on or before July 22, 1999. MANUFACTURER Steese's Ice Cream Company, Grove City, Pennsylvania. RECALLED BY Manufacturer, by telephone on July 23, 1999, by press release on July 24, 1999. Firm-initiated recall complete. DISTRIBUTION Pennsylvania. QUANTITY Undetermined. REASON Products do not bear complete ingredient statements and contain the following undeclared ingredients: a) eggs and wheat; b-e) cashews; f-h) pecans; i) walnuts; j-l) peanuts.

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS II

________ PRODUCT OVA brand Orange Peach Drink, in 473 ml glass bottles, a non- carbonated beverage. Recall #F-010-0. CODE All codes. MANUFACTURER Grayco Sales, Ltd., Ontario, Canada. RECALLED BY Consolidated Stores Corporation, Columbus, Ohio, by E-mail on August 25, 1999. Completed recall resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets. DISTRIBUTION Nationwide. QUANTITY 37,536 bottles were distributed. REASON Product contains the uncertified color additives tartrazine (certifiable as FD&C Yellow No. 5) and sunset yellow (certifiable as FD&C Yellow No. 6). ________ PRODUCT Life Savers Candy, in 5.2 ounce (150g) plastic bags. Recall #F-011-0. CODE All codes. MANUFACTURER Beta Brands Ltd., Ontario, Canada. RECALLED BY Consolidated Stores Corporation, Columbus, Ohio, by E-mail, on August 25, 1999. Completed recall resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets. DISTRIBUTION Nationwide. QUANTITY 72,000 bags were distributed. REASON The product contains the uncertified color additives tartrazine (certifiable as FD&C Yellow No. 5), sunset yellow (certifiable as FD&C Yellow No. 6), and brilliant blue FCF (certifiable as FD&C Blue No. 1). It also contains an unapproved color formerly certifiable as FD&C Red. No. 2. ________ PRODUCT Ice Cream and Sherbet as follows: a) Butter Almond Ice Cream, in 2 1/2 gallon size containers b) Coconut Cream Ice Cream in 1/2 gallon and 2 1/2 gallon size containers c) Butterscotch Ice Cream in 1/2 gallon and 2 1/2 gallon size containers d) Banana Cream Ice Cream in 2 1/2 gallon size containers e) Bubble Gum Ice Cream in 2 1/2 gallon size containers f) Cookies and Cream Ice Cream in 1/2 gallon and 2 1/2 gallon size containers g) Graham Cracker Ice Cream in 2 1/2 gallon size containers h) Peppermint Stick Ice Cream in 1/2 gallon and 2 1/2 gallon size containers i) Sunrise (Orange Sherbet and Vanilla Ice Cream) in 2 1/2 gallon size containers j) Orange Sherbet in 1/2 gallon and 2 1/2 gallon size containers k) Orange Pineapple Ice Cream in 1/2 gallon size containers. Recall #F-027/037-0. CODE Products are not coded. All products distributed on or before July 22, 1999. MANUFACTURER Steese's Ice Cream Company, Grove City, Pennsylvania. RECALLED BY Manufacturer, by telephone on July 23, 1999, by press release on July 24, 1999. Firm-initiated recall complete. DISTRIBUTION Pennsylvania. QUANTITY Undetermined. REASON Products do not bear complete ingredient statements and contain the following undeclared ingredients: a) almonds; b-e) FD&C Yellow No. 5, and FD&C Yellow No. 6; f-g) FD&C Yellow No. 5 and wheat; h) FD&C Yellow No. 5; i-k) FD&C Yellow No. 6. ________ PRODUCT Prawn Crackers (Color) Banh Phong Tom Mau in 8 ounce boxes packed 60 per case and 16 ounce boxes packed 50 per case. Recall #F-043-0. CODE None. All lots distributed in 1999. MANUFACTURER Packed for Guangdong Foods I & E Company, Ltd., Guangdong, China. RECALLED BY Great Union Wholesale, Inc., Houston, Texas, by letter dated September 23, 1999. Firm-initiated recall ongoing. DISTRIBUTION Texas and Florida. QUANTITY 400 8-ounce boxes and 120 16-ounce boxes were distributed REASON The product contains the unapproved color additives Ponceau 4R and azorubine. In addition it contains the uncertified colors sunset yellow FCF (certifiable as FD&C Yellow No. 6), brilliant blue FCF (certifiable as FD&C Blue No. 1), and tartrazine (certifiable as FD&C Yellow No. 5).

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS III

________ PRODUCT Shaw's Tender Blue Lake French Style Green Beans, in 14.5 ounce cans. Recall #F-014-0. CODE 1307G 89161. MANUFACTURER Chiquita Brands (formerly KNA Friday Canning), Gillett, Wisconsin. RECALLED BY Shaw's Supermarkets, Inc., East Bridgewater, Massachusetts, by letter on August 26, 1999, October 16 and 18, 1999, and by press release on October 22, 1999. Firm-initiated recall ongoing. DISTRIBUTION Connecticut, Massachusetts, Maine, New Hampshire, Rhode Island, Vermont. QUANTITY 1,326 cases (24 cans per case) were distributed. REASON Product may be contaminated with filth. ________ PRODUCT Ice Cream as follows: a) Black Walnut Ice Cream in 1/2 gallon size containers b) Butter Pecan Ice Cream in 1/2 gallon size containers c) Chocolate Peanut Butter Ripple Ice Cream in 1/2 gallon size containers d) Peanut Butter Ripple Ice Cream in 1/2 gallon size containers. Recall #F-038/041-0. CODE Products are not coded. All products distributed on or before July 22, 1999. MANUFACTURER Steese's Ice Cream Company, Grove City, Pennsylvania. RECALLED BY Manufacturer, by telephone on July 23, 1999, by press release on July 24, 1999. Firm-initiated recall complete. DISTRIBUTION Pennsylvania. QUANTITY Undetermined. REASON Products are misbranded because the labeling lists "nuts" as an ingredient but does not specify which nut is in the product. ________ PRODUCT Hunt's Ketchup Plastic Squeeze Bottles, sizes 24 ounces, 32 ounces and 50 ounces. Recall #F-044-0. CODE The following codes and dates are affected: CODE DATE R95A 05/10/99 R95B 05/11/99 R95C 05/12/99 R95D 05/13/99 R95E 05/14/99 R95F 05/15/99 R95H 05/16/99 R95J 05/17/99 R95K 05/18/99 R95L 05/19/99 R95M 05/20/99 R95N 05/21/99 R95O 05/22/99 R95P 05/23/99 R95R 05/24/99 R95S 05/25/99 R95T 05/26/99 R95U 05/27/99 R95W 05/28/99. MANUFACTURER ConAgra Grocery Products Cannery, Perrysburg, Ohio. RECALLED BY ConAgra Grocery Products Cannery, Fullerton, California, by telephone on June 11, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 294,419 cases were distributed. REASON Product is contaminated with Lactobacillus fermentum and Lactobacillus cellobiosus.

RECALLS AND FIELD CORRECTIONS: COSMETICS -- CLASS II

________ PRODUCT Liquid Mouth Rinse/Mouth Wash, alcohol free, packed in 4 fluid ounce plastic bottles, under the BeFresh and MediFresh labels. Recall F-045-0. CODE Lot #08309-B. MANUFACTURER Chester Labs, Inc., Cincinnati, Ohio. RECALLED BY Manufacturer, by telephone on October 19, 1999, and by letter on October 26, 1999. Firm-initiated recall ongoing. DISTRIBUTION Florida, New York, Tennessee, Ohio, Illinois, Texas. QUANTITY 444 cases (26,640 bottles) were distributed; firm estimated that 170 cases remained on market at time of recall initiation. REASON Product is contaminated with Enterobacter gergoviae.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II

________ PRODUCT Red Blood Cells, Autologous. Recall #B-038-0. CODE Unit #J43291. MANUFACTURER Blood Assurance, Inc., Chattanooga, Tennessee. RECALLED BY Manufacturer, by telephone on August 31, 1999. Firm-initiated recall complete. DISTRIBUTION Georgia. QUANTITY 1 unit was distributed. REASON Blood product tested initially reactive for anti-HCV, and negative by a single repeat test. ________ PRODUCT a) Red Blood Cells; b) Platelets; c) Fresh Frozen Plasma; d) Cryoprecipitated AHF; e) Platelets for Use in Manufacturing Non-Injectable Products; f) Recovered Plasma Liquid For Manufacturing Non-Injectable Products; g) Recovered Plasma. Recall #B-065/071-0. CODE Unit Numbers: a) 31507, 57170, 99944, 118635, 173081, 194857, 285506, 292734, 314544, 354293, 462826, 507843, 0562387, 0594245, 0645267 b) 57170, 99944, 118635, 173081, 194857, 292734, 354293, 0562387; c) 173081, 462826; d) 9005621 e) 507843, 0594245; f) 114331, 14425, 46007, 48744, 90708, 92982, 477357; g) 507843, 31507, 57170, 99944, 194857, 245265, 247572, 285506, 292734, 314544, 354293, 0562387, 0594245, 0645267, 9005621. MANUFACTURER Memorial Blood Centers of Minnesota, Minneapolis, Minnesota. RECALLED BY Manufacturer, by letters dated June 15, 22, or 28, 1999. Firm- initiated recall complete. DISTRIBUTION Minnesota, North Carolina, Ohio, Virginia, California, Florida, Texas, Pennsylvania, Switzerland. QUANTITY a) 15 units; b) 8 units; c) 2 units; d) 1 unit; e) 2 units; f) 7 units; g) 15 units were distributed. REASON Blood products tested negative for the antibody to the human immunodeficiency virus type 1 (anti-HIV-1), but were collected from donors who previously tested repeatedly reactive for anti- HIV-1, Western blot indeterminate. ________ PRODUCT a) Red Blood Cells; b) Platelets; c) Fresh Frozen Plasma; d) Cryoprecipitated AHF; e) Recovered Plasma Liquid For Non-Injectable Products; f) Recovered Plasma, Pooled. Recall #B-076/081-0. CODE Unit Numbers: a) 74271, 163310, 268929, 399080, 92229, 107723; b) 74271, 163310, 268929, 92229, 107723; c) 74271, 268929, 92229, 107723; d) 9007428; e) 399080; f) 9007428. MANUFACTURER Memorial Blood Centers of Minnesota, Minneapolis, Minnesota. RECALLED BY Manufacturer, by letters dated June 15, 21, or 28, 1999. Firm- initiated recall complete. DISTRIBUTION Minnesota, California, Florida. QUANTITY a) 6 units; b) 5 units; c) 4 units; d) 1 unit; e) 1 unit; f) 1 unit distributed. REASON Blood products either tested initially reactive for the antibody to the human immunodeficiency virus type 1 (anti-HIV-1), or which tested negative for the antibody to the human immunodeficiency virus, but were collected from donors who previously tested repeatedly reactive for either anti-HIV-1 or anti HIV-1/2. ________ PRODUCT Blood Collection Systems. Recall #B-084-0. CODE 50 lot numbers of products manufactured from May 04 through July 05, 1999. MANUFACTURER Medsep Corporation, a subsidiary of Pall Corporation, Covina, California. RECALLED BY Manufacturer, by letter dated August 2, 1999, followed by telephone. Firm-initiated recall complete. DISTRIBUTION Nationwide and England. QUANTITY 390,076 units were distributed. REASON Blood collection bags contained low volumes of anticoagulant. ________ PRODUCT Red Blood Cells, Leukocytes Reduced, Irradiated; b) Recovered Plasma. Recall #B-086/087-0. CODE Unit #40GE37390. MANUFACTURER American Red Cross Blood Services, Peoria, Illinois. RECALLED BY Manufacturer, by letter dated July 23, 1999. Firm-initiated recall complete. DISTRIBUTION Illinois and California. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor taking the drug Proscar. ________ PRODUCT Salvage Plasma. Recall #B-089-0. CODE Unit #7023525. MANUFACTURER Department of the Air Force, Keesler AFB, Mississippi. RECALLED BY Manufacturer, by telephone on September 12, 1997. Firm-initiated recall complete. DISTRIBUTION Florida. QUANTITY 1 unit was distributed. REASON Blood product was not tested for the antibodies to the human immunodeficiency virus type1/2, and Hepatitis B surface antigen (HBsAG), Hepatitis C virus encoded antigen (anti-HCV), Hepatitis B core antigen (anti-HBc), and syphilis. ________ PRODUCT a) Red Blood Cells; b) Platelets; c) Platelets, Pheresis; d) Fresh Frozen Plasma; e) Recovered Plasma. Recall #B-100/104-0. CODE FDA, Center for Biologics Evaluation and Research, Contact Office of Compliance (301) 827-6220 for individual unit numbers recalled. MANUFACTURER Blood Systems, Inc., Tupelo, Mississippi. RECALLED BY Blood Systems, Inc., Scottsdale, Arizona, by telephone on November 21 and 26, 1997, by letter and fax on January 30, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama, California, Illinois, Louisiana, Mississippi, New Jersey, New York, Tennessee. QUANTITY a) 111 units; b) 10 units; c) 14 units; d) 19 units; e) 82 units were distributed. REASON Blood products were collected from donors whose suitability was not adequately determined. ________ PRODUCT a) Red Blood Cells; b) Recovered Plasma. Recall #B-105/106-0. CODE Unit #19233-2718. MANUFACTURER United Blood Services, Reno, Nevada. RECALLED BY Blood Systems, Inc., Scottsdale, Arizona, by telephone or letter on February 6, 1998. Firm-initiated recall complete. DISTRIBUTION Texas and Switzerland. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor who received weekly injections of gold salt. ________ PRODUCT a) Red Blood Cells; b) Platelets. Recall #B-120/121-0. CODE Unit Numbers: 22KV05788, 22GE46099. MANUFACTURER American National Red Cross, Philadelphia, Pennsylvania. RECALLED BY Manufacturer, by letters dated October 13 and 30, 1998, February 3, 1998, and March 1, 1999. Firm-initiated recall complete. DISTRIBUTION Pennsylvania and New Jersey. QUANTITY 2 unit of each component were distributed REASON Blood products were collected from donors who reported having lived or traveled in area designated as endemic for malaria. ________ PRODUCT a) Red Blood Cells, Leukoreduced; b) Recovered Plasma. Recall #B-123/124-0. CODE Unit #03GV11621. MANUFACTURER American Red Cross Blood Services, Atlanta, Georgia. RECALLED BY Manufacturer, by letter dated October 28, 1998. Firm-initiated recall complete. DISTRIBUTION Georgia and California. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a unsuitable donor. ________ PRODUCT Platelets, Pheresis, Leukocytes Reduced, Irradiated. Recall #B-125-0. CODE Unit #0328384-A2. MANUFACTURER Inland Northwest Blood Center, Spokane, Washington. RECALLED BY Manufacturer, by telephone on July 22, 1999, and by letter dated July 27, 1999. Firm-initiated recall complete. DISTRIBUTION Washington state. QUANTITY 1 unit was distributed. REASON Blood product had incomplete documentation of irradiation.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS III

________ PRODUCT Reverse Group 4TM, Pooled Cells for Serum Confirmation of ABO Group; b) Screening Cell PoolTM, for use in detection of unexpected antibodies. Recall #B-082/083-0. CODE a) Lot 0629 EXP 04 Aug 99; b) Lot 0629, EXP 04 Aug 99. MANUFACTURER Gamma Biologicals, Inc., Houston, Texas. RECALLED BY Manufacturer, by letter dated July 6, 1999. Firm-initiated recall complete. DISTRIBUTION Nationwide, Japan, Netherlands, Saudi Arabia. QUANTITY a) 198 vials; b) 226 vials were distributed. REASON In-vitro diagnostic products were contaminated with Serratia marcesans. ________ PRODUCT Red Blood Cells, Irradiated, Leukocytes Reduced. Recall #B-085-0. CODE Unit #X46547. MANUFACTURER Aurora Area Blood Bank, Aurora, Illinois. RECALLED BY Manufacturer, by telephone on August 19, 1999. Firm-initiated recall complete. DISTRIBUTION Illinois. QUANTITY 1 unit was distributed. REASON Blood product was labeled with an extended expiration date. ________ PRODUCT Sandoglobulin IV, Immune Globulin Intravenous (Human), in 12 gram vials, indicated for the maintenance treatment of patients with primary immunodeficiencies. Recall #B-088-0. CODE Lot number on box (lot number on vial): #561Z4409 (7.409.561.0), 029A6093 (8.409.208.0), 047A6478 (8.409.362.0), 053A6697 (8.409.433.0), 053A6698 (8.409.433.0), 072A8945 (8.409.599.0), 080B9812 (9.409.229.0), 085B9813 (9.401.255.0), 090B0161 (9.401.292.0), 561Z4098 (7.409.561.0), 072A8946 (8.409.599.0). MANUFACTURER ZLB Central Laboratory, Bern, Switzerland. RECALLED BY Novartis, East Hanover, New Jersey, by fax or letter dated August 27, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 24,585 vials were distributed. REASON Eleven lots of Immune Globulin Intravenous (Human) exceeded the residual moisture specification. ________ PRODUCT Recovered Plasma. Recall #B-131-0. CODE Unit #2549185. MANUFACTURER Department of Army, Armed Services Blood Bank Center, Tacoma, Washington. RECALLED BY Manufacturer, by letters dated February 10, 1999, and May 28, 1999. Firm-initiated recall complete. DISTRIBUTION Pennsylvania. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who reported taking the drug mesalamine.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

________ PRODUCT RUBELLA IgG ELISA TEST KIT, diagnostic test kit used for detection of IgG antibody to rubella virus in human serum. Recall #Z-085-0. CODE Catalog/Product Number: RUE100-E, Lot #9DMGWW, EXP 4/30/00. MANUFACTURER Gull Laboratories, Salt Lake City, Utah. RECALLED BY Meridian Diagnostics, Inc., Cincinnati, Ohio, by telephone beginning August 24, 1999, and by fax between August 24 and 27, 1999. Firm-initiated recall complete. DISTRIBUTION California, Florida, Massachusetts, Michigan, Minnesota, Missouri, New York, Oklahoma, Pennsylvania, Texas. QUANTITY 178 units. REASON Optical density (O.D.) values for the reference serum are outside of acceptable limits. ________ PRODUCT LIFEPAK 500 Automated External Defibrillator designed to be used by first responders to cardiac emergencies. Recall #Z-086-0. CODE Part Numbers: 3005400-000, 001, 002, 003, 004, 005, 009. 010, 011, 012, 013, 014, 018, 020, 021, 022, 029, 030, 031, 064, 065, 066, 067, 068, 070, 071, 074, 076, 078, 082, 083, 084 and 085. Serial Numbers: All devices distributed prior to 8/21/98. The serial number range is for selected 8 digit serial numbers less than 10486696. MANUFACTURER Medtronic Physio-Control Corporation, Redmond, Washington. RECALLED BY Manufacturer, by visit on or about September 24, 1999. Firm- initiated field correction ongoing. DISTRIBUTION Nationwide and international. QUANTITY 10,746 units were distributed. REASON Excessive electrical noise interfering with ECG analysis that, when present, may cause the shock advisory algorithm to render an incorrect decision. ________ PRODUCT Blood Pressure Cuff for the BP 410 and 412 Monitor: a) Part 17402-001: Blood pressure cuff, child Part 17403-001: Blood pressure cuff, adult Part 17403-003: Blood pressure cuff, large adult b) Part 00304-001: 412 Blood pressure monitor (includes 71403- 001); Part 00304-002: 412 Blood pressure monitor 50hz (includes 17403-001); Part 00304-B-001: 412 Blood pressure monitor 50hz (includes 17403-001). Recall #Z-093/094-0. CODE None (n/a). MANUFACTURER Colin Medical Instruments Corporation, San Antonio, Texas. RECALLED BY Quinton Instrument Company, Bothell, Washington, by letter on or about October 15, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 630 units were distributed. REASON The bladder and air hose of the blood pressure cuff contain undeclared rubber latex. ________ PRODUCT Portable Liquid Oxygen Units Models: AA-1000-EA, AA-1005-AA, AA- 1010-AA, AB-1001-EA, AB-1002-EC, AB-1005-AA, AB-1005-AE, AB-1005- AG, AB-1005-EA, AB-1015-EA, AC-1005-AA, AD-1003-EA, AD-1005-AA, AD-1012-AA, AD-1017-EA, and AD-1014-EB. Recall #Z-095/111-0. CODE No Lot Numbers. MANUFACTURER Penox Technologies, Inc., Pittston, Pennsylvania. RECALLED BY Manufacturer, by letter sent on August 12, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 1,144 units were distributed. REASON Portable liquid oxygen unit was distributed with unrestricted flow regulator. ________ PRODUCT Schuhli Implant System, designed to lock a screw to a plate, independent of bony contact with the plate: Part numbers 219.401, 219.401S, 219.402, 219.402S, 219.404, 219.404S, 219.411, 219.411S, 219.412, 219.412S, 219.414, 219.414S, 419.401, 419.401S, 419.402, 419.402S, 419.404, 419.404S, 419.411, 419.411S, 419.412, 419.412S, 419.414, and 419.414S. Recall #Z- 112/117-0. CODE All lots are involved. MANUFACTURER Synthes (USA), Monument, Colorado. RECALLED BY Synthes (USA), Paoli, Pennsylvania, by E-mail on October 30, 1997. Firm-initiated recall complete. DISTRIBUTION Nationwide, Canada, Switzerland. QUANTITY 3,827 units. REASON In some instances the device may not lock securely into the plate. ________ PRODUCT Quickie Power Wheelchairs P200 Series, a powered wheelchair; battery operated medical device with wheels that is intended for medical purpose to provide mobility to persons restricted to a sitting position: a) Model P120; b) Model P200; c) Model P210; d) Model P300; e) Model P320. Recall #Z-118/122-0. CODE Model P120, Serial #s: P19-0290 to P19-6340 Model P200, Serial #s: P2-1006 to P2-13486 Model P210, Serial #s: P2-10010 to P2-13517 Model P300, Serial #s: P3-10500 to P3-12197 Model P320, Serial #s: P32-1000 to P32-4420. MANUFACTURER Sunrise Medical Quickie Designs, Inc., Fresno, California. RECALLED BY Manufacturer, by letter on October 26, 1999, followed by telephone. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 11,030 units. REASON The failure of the front two bolts-caster plate-can result in the caster assembly separating from the chair frame. ________ PRODUCT Sterngold Black Era-RV (Reduced Vertical) Males Intended for the retention of partial dentures: Product Numbers: a) 811105 Era RV Starter Kit b) 811115 Era RV Attachment c) 811125 Era RV Black Males d) 811166 Assorted Era RV Males e) 811168 Era RV Housing f) 811203 Era Offset RV 4.5mm g) 811204 Era Offset RV 2.5mm. Recall #Z-123/129-0. CODE Lot Numbers: a) W1579; b) W1262, W1263, W1305, W1293, W1413, W1305,W1340, W1452, W1475, W1476,W1508, W1565, W1608, W1639, W1655, W1780, W1795, W1795, W1812, W1841, 990901-1, 990902-1; c) W1178, W1483; d) W1302, W1774 e) W1176, W1214, W1458; f) W1277, W1283, W1319, W1705 g) W1278, W1284, W1318, W1660, W1767. MANUFACTURER Spectrum Plastics, Assonia, Connecticut. RECALLED BY Sterngold, Attleboro, Massachusetts, by letter dated September 28, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 8,037 units. REASON Incorrect height dimensions to the denture attachment. ________ PRODUCT LSL Sterile Tracheotomy Care and Clean Set, sterile, single patient use tray. Recall #Z-131-0. CODE Product #LSL5110, Lot #9G3020. MANUFACTURER LSL Industries, Inc., Chicago, Illinois (kit); Jiang Su Jin Hong Corporation, Jintan City, Jiangsu, China (gauze). RECALLED BY LSL Industries, Inc., Chicago, Illinois, by letter dated September 23, 1999. Firm-initiated recall ongoing. DISTRIBUTION Maine, Illinois, Missouri, New Jersey, Puerto Rico. QUANTITY 1,160 trays. REASON The kits may contain a contaminated gauze component. The gauze may contain an ETO resistant mold. ________ PRODUCT FreeHand Implantable Epimysial Electrodes for the NeuroControl FreeHand System, Part #1320-A and Model 200-19, all sizes/lengths. The system uses electrical muscle stimulation which is intended to improve a patient's ability to grasp, hold, and release objects. Recall #Z-132-0. CODE The affected serial numbers are: 243, 253, 259, 263, 289, 290, 368, 471, 493, 512, 800, 804, 820, 842, 850, 852, 873, 885, 900, 914, 893, 955, 968, 971, 977, 980, 983, 1077, 1127, 1151, 1153, 1172, 1217, 1221, 1241, 1250, 1255, 1261, 1307, 1564, 1582, 1604, 1606, 1630, 1635, 1651, 1955, 1965, 1966, 1976, 1981, 1999, 2001, 2006, 2044, 2045, 2046, 2055, 2325, 2062, 2232, 2063, 2065, 2067, 2008, 2069, 2160, 2198, 2229, 2234, 2235, 2236, 2237, 2249, 2253, 2276, 2291, 2293, 2294, 2295, 2299, 2315, 2323, 2326, 2327, 2331, 2424, and 2425. MANUFACTURER NeuroControl Corporation, Valley View, Ohio (kits); Bio-Control Technology, Inc., Indiana, Pennsylvania (electrodes). RECALLED BY NeuroControl Corporation, Valley View, Ohio, by telephone beginning on November 13, 1998. Firm-initiated recall ongoing. DISTRIBUTION New York, Pennsylvania, Indiana, Australia, United Kingdom, France. QUANTITY 87 electrodes included in a total of 6 kits. REASON Lack of assurance that the electrodes were sterilized. ________ PRODUCT a) Percutaneous Transluminal Angioplasty (PTA) Catheter, Catalog numbers: PTA-OS253, 303 and 403 b) Z-MED Dilation Catheter, Catalog Numbers: PDZ-S c) Z-MED II Dilatation Catheter, Catalog Numbers: PDS-630, 643 and 655 d) PTA-OS Balloon Catheter, Catalog Numbers: VSB-030 and 040 e) Ghost II Dilation Catheter, Catalog Numbers: GII-109, 113, 121, 129, 130, 137, 138, 141. Recall #Z-135/139-0. CODE a) OS-0013THRU OS-0014, AND OS-0022 THRU OS-0024 b) Z-0774 c) ZZ-0240 THRU ZZ-0242 d) OS-0025/0026 e) GG0138 THRU GG0145. MANUFACTURER NUMED, Inc., Hopkinton, New York. RECALLED BY Manufacturer, by fax between January 29, 1999 and February 5, 1999. Firm-initiated recall complete. DISTRIBUTION Massachusetts, United Kingdom, South Africa, Argentina. QUANTITY 302 catheters were distributed. REASON Non-sterility. ________ PRODUCT Surface Smoothness Sensor, Model 4206. Recall #Z-141-0. CODE Model #4206. MANUFACTURER Honeywell-Measurex Corporation, Cupertino, California. RECALLED BY Manufacturer. FDA Approved the firm's corrective action plan on May 11, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide. QUANTITY 14 units. REASON The laser product failed to comply with the Federal laser product performance standard, 21 CFR 10 1040.10(d) in that the product permitted human access to laser radiation in excess of the product class, and 1040.10(f)(2) in that the product failed to incorporate a safety interlock that was fail safe or redundant. ________ PRODUCT Abbott TestPack Plus H. pylori Reaction Discs, in vitro diagnostic, rapid, visually read, qualitative immunoassay for the detection of human IgG antibodies specific to Helicobacter pylori in serum, plasma and whole blood: List 1D04-05 - 5 test kit (whole blood); List 1D04-16 - 20 test kit (whole blood); List 6C71-16 - 20 test kit (whole blood, plasma or serum). Recall #Z-144/145-0. CODE List number 1D04-05: Lot numbers 54412M300, 56628M200 List number 1D04-16: Lot numbers 54334M300, 56629M200 List number 6C71-16: Lot numbers 54417M300, 55211M300, 56631M200. MANUFACTURER Abbott Laboratories, Inc., North Chicago, Illinois. RECALLED BY Abbott Laboratories, Inc., Abbott Park, Illinois, by letter dated October 22, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 5,003 kits were distributed. REASON False positive results with some negative samples. ________ PRODUCT Abbott Gonozyme Diagnostic Kit, an EIA for detection of Neisseria gonorrhoeae antigens in urethral and endocervical swab specimens from humans; 100 test kit: List 6502-24 - Kit with female swabs List 6502-25 - Kit with male swabs List 6502-62 - Reagent Kit with no swabs. Recall #Z-148/150-0. CODE Lot number 49762M300. MANUFACTURER Abbott Laboratories, Inc., Abbott Park, Illinois. RECALLED BY Manufacturer, by letter dated October 19, 1999. Firm-initiated recall complete. DISTRIBUTION Texas, Germany, Italy, Japan, Singapore, Taiwan. QUANTITY 136 kits were distributed. REASON Stability failure. ________ PRODUCT Ameda Disposable Vacuum Assist Cup, a sterile, single use device for use with electric or hand-held vacuum pumps to facilitate vaginal delivery of term or near term infants exhibiting fetal distress. Recall #Z-151-0. CODE Sotck #17420, Lot #9E10. MANUFACTURER Hollister, Inc., Cary, Illinois. RECALLED BY Hollister, Inc., Libertyville, Illinois, by letter dated October 22, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 1,780 units were distributed. REASON Some of the tubing adapters do not have the black "O-rings" in place. ________ PRODUCT Alpha IQ Mammographic X-Ray Units: a) Model IQ; b) Model RT. Recall #Z-152/153-0. CODE Serial numbers 8140-8268, except 8251, 8189, 8254, 8257, 8219, and 8238. MANUFACTURER Instrumentarium Corporation, Imaging Division, Tuusula, Finland. RECALLED BY Instrumentarium Imaging, Inc., Milwaukee, Wisconsin, by field modification instructions dated October 28, 1999. Firm-initiated field correction ongoing. Nationwide. QUANTITY 62 units were distributed. REASON There can be unintended motion of the C-arms of the X-ray units due to insufficient switch insulation.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III

________ PRODUCT Immunocard Toxin A, a diagnostic test kit. Recall #Z-133-0. CODE Catalog # 711050, Lot Numbers: #711050.112 and #711050.113. Both lots expire on 5/29/00. MANUFACTURER Meridian Diagnostics, Inc., Cincinnati, Ohio. RECALLED BY Manufacturer, by telephone on beginning on August 8, 1999, and by letters mailed on August 8, 1999, and September 17, 1999. Firm- initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 291 kits were distributed. REASON Kits contain an improper sample transfer pipet (50ul instead of 25 ul). ________ PRODUCT Premier Giardia, a diagnostic test kit. Recall #Z-134-0. CODE Catalog #614096, Lot Numbers: 614096.010 EXP 9/8/99 and 614096.011 EXP 11/10/99. MANUFACTURER Meridian Diagnostics, Inc., Cincinnati, Ohio. RECALLED BY Manufacturer, by letter dated October 13, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 601 kits were distributed. REASON Weak positive test samples may be reported as false negative. ________ PRODUCT Abbott TestPack Chlamydia 20T-CE Kit, in-vitro diagnostic enzyme immunoassay for the detection of Chlamydia trachomatis antigen from endocervical swab specimens; 20 tests per kit. Recall #Z- 140-0. STRONG>CODE List 03436-18; Kit lot #56934M200; Swab lot #170. MANUFACTURER Abbott Laboratories, Inc., North Chicago, Illinois. RECALLED BY Abbott Laboratories, Inc., Abbott Park, Illinois, by letter dated October 19, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 48 kits were distributed. REASON Swab component expires prior to the kit expiration date. ________ PRODUCT ARCHITECT Estradiol Reagent Kit: List 6C22-20 and List 6C22-25, in-vitro diagnostic Chemiluminescent Microparticle Immunoassay (CMIA) for the quantitative determination of estradiol in human serum and plasma. Z-142/143-0. CODE List 6C22-20 (400 tests): lots 47955M300, 54859M100, 54859M103, 54965M300, 55686M200; List 6C22-25 (100 tests): lots 47955M301, 54859M101, 54965M301, 55686M201. MANUFACTURER Abbott Laboratories, Inc., Abbott Park, Illinois. RECALLED BY MANUFACTURER, by telephone on October 20, 1999, and by letters dated October 19, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 825 400-test kits; 463 100-test kits were distributed. REASON Under recovery of estradiol with undiluted patient specimens. ________ PRODUCT Abbott Test Pack Plus hCG Combo with on Board Controls (OBC) Reaction Disc, a self-performing immunoassay designed for the qualitative detection of human chorionic gonadotropin (hCG) in serum and urine for early detection of pregnancy. Recall #Z-146- 0. CODE List 7B34, Lot Numbers 52412M200, 52414M200, 53867M100, 55058M200. MANUFACTURER Abbott Laboratories, Inc., North Chicago, Illinois. RECALLED BY Abbott Laboratories, Inc., Abbott Park, Illinois, by letter dated October 22, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 32,036 kits were distributed. REASON Lots may produce false positive results. ________ PRODUCT Abbott Vision Chem Controls, multianalyte controls used for verifying the performance of specific chemistry tests on the Abbott Vision System. Recall #Z-147-0. CODE< List 1433-02, Lot Numbers 51719M405, 51719M406, 51719M407. MANUFACTURER Abbott Laboratories, Inc. North Chicago, Illinois. RECALLED BY Abbott Laboratories, Inc., Abbott Park, Illinois, by letter dated October 19, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide and international. QUANTITY 2,589 kits were distributed. REASON Chem Control II running below specification range when used with SGPT assay. END OF ENFORCEMENT REPORT FOR NOVEMBER 10, 1999.

[FDA HOME PAGE]