The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

June 2, 1999                                          99-22

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS I

________ PRODUCT Rich Flower brand Dried Sweet Potato, in 7 ounce rigid plastic bags, product of China. Recall #F-437-9. CODE None. MANUFACTURER Garden Food (Shenzhen) Company, Ltd., China (packer). RECALLED BY CL Import Company, Inc., Brooklyn, New York (importer), by telephone on October 19, 1998. Completed field correction (relabeling) resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets. DISTRIBUTION New York. QUANTITY 2 cases were distributed. REASON Product contained undeclared sulfites. ________ PRODUCT Italian Sun Dried Tomatoes, in 16 ounce cryovac, plastic bags. Recall #F-440-9. CODE None. MANUFACTURER Undetermined. RECALLED BY Mrs. Mazzula's Food Products, Inc., Edison, New Jersey (repacker), by press release on April 1, 1999, and by letter on April 9, 1999. Ongoing field correction (relabeling) resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets. DISTRIBUTION New Jersey. QUANTITY 110 cartons (12 bags per case) were distributed. REASON Product contains undeclared sulfites.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS II

________ PRODUCT a) Red Blood Cells; b) Fresh Frozen Plasma; c) Recovered Plasma. Recall #B-719/721-9. CODE Unit numbers: a) 28LN13640, 28LQ15071; b) 28LQ15071; c) 28LN13640, 28LX04404. MANUFACTURER American Red Cross Blood Services, Charelston, South Carolina. RECALLED BY American Red Cross Blood Services, Savannah, Georgia, by letter or fax on either February 17 or 18, 1998, or by telephone on February 16, 1998. Firm-initiated recall complete. DISTRIBUTION Georgia, South Carolina, Puerto Rico, California, Switzerland. QUANTITY a) 2 units; b) 1 unit; c) 2 units were distributed. REASON Blood products were collected from a donor with a history of hepatitis. ________ PRODUCT Red Blood Cells. Recall #B-793-9. CODE Unit #41LX27263. MANUFACTURER The American National Red Cross, Birmingham, Alabama. RECALLED BY Manufactaurer, by letter dated September 2, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who reported travel to an area designated as endemic for malaria. ________ PRODUCT Red Blood Cells. Recall #B-794-9. CODE Unit #41FQ57714. MANUFACTURER The American National Red Cross, Birmingham, Alabama. RECALLED BY Manufacturer, by letter dated October 5, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who reported travel to an area designated as endemic for malaria. ________ PRODUCT a) Red Blood Cells; b) Platelets; c) Fresh Frozen Plasma. Recall #B-798/800-9. CODE Unit Numbers: a) 11-147280; b) 11-147280, 11-159274; c) 11-147280, 11-159274. MANUFACTURER United Blood Services, El Paso, Texas. RECALLED BY Manufacturer, by letter dated February 3, 1999. Firm-initiated recall complete. DISTRIBUTION Texas and New York. QUANTITY a) 1 unit; b) 2 units; c) 2 units were distributed. REASON Blood products tested negative for HBsAg, but were collected from a donor who previously tested repeatedly reactive for HbsAg. ________ PRODUCT Platelets Pheresis. Recall #B-801-9. CODE Unit numbers: 049FP06220, 049FP06493, 049FP06508, 049GP00525, 049GP00671, 049GP00713. MANUFACTURER American Red Cross Blood Services, Waco, Texas. RECALLED BY American Red Cross Blood Services, Tulsa, Oklahoma, by letters dated January 20, 1999, and February 19, 1999. Firm-initiated recall complete. DISTRIBUTION Texas. QUANTITY 6 units were distributed. REASON Blood products had unacceptable platelet counts. ________ PRODUCT a) Red Blood Cells; b) Platelets. Recall #B-803/804-9. CODE Unit numbers: 5102179 and 5084896. MANUFACTURER BloodCare, doing business as Carter BloodCare, Dallas, Texas. RECALLED BY Manufacturer, by letter dated December 10, 1998. Firm-initiated recall complete. DISTRIBUTION Texas. QUANTITY 2 units of each component were distributed. REASON Blood products were collected from a donor who reported travel to an area designated as endemic for malaria. ________ PRODUCT a) Red Blood Cells; b) Platelets. Recall #B-805/806-9. CODE Unit numbers: a) 11315-1160, 11316-4855, 11317-0095, 11317-6777 b) 11315-1160, 11316-4855. MANUFACTURER United Blood Services, El Paso, Texas. RECALLED BY Blood Systems, Inc., Scottsdale, Arizona, by letter dated January 26, 1999. Firm-initiated recall complete. DISTRIBUTION New Mexico and Texas. QUANTITY a) 4 units; b) 2 units were distributed. REASON Blood products were collected from a donor who reported travel to an area designated as endemic for malaria. ________ PRODUCT Platelets, Pheresis. Recall #B-809-9. CODE Unit #49P42789. MANUFACTURER American Red Cross Blood Services, Tulsa, Oklahoma. RECALLED BY Manufacturer, by telephone on January 31, 1999. Firm-initiated recall complete. DISTRIBUTION Oklahoma. QUANTITY 1 unit was distributed. REASON The first half of a split Platelets, Pheresis unit, implicated in a transfusion reaction, was found to be contaminated with Staphylococcus aureus. The second half of the unit was distributed. ________ PRODUCT Source Plasma. Recall #B-810-9. CODE Unit 98YARC7412 (Donor Id. A00576590). MANUFACTURER Community Bio-Resources, Inc., Fayetteville, Arkansas. RECALLED BY Community Bio-Resources, Inc., Hoover, Alabama, by fax dated December 4, 1998. Firm-initiated recall complete. DISTRIBUTION Austria. QUANTITY 1 unit was distributed. REASON Blood product was collected from a donor who had received an MMR vaccination within four weeks of donation. _______ PRODUCT a) Red Blood Cells; b) Platelets, Pheresis. Recall #B-811/812-9. CODE Unit numbers: a) 55R15127; b) 55FQ00336. MANUFACTURER American Red Cross Blood Services, Little Rock, Arkansas. RECALLED BY Manufacturer, by letter dated April 8, 1999. Firm-initiated recall complete. DISTRIBUTION Arkansas. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor who had sexual contact with someone who had lived in an HIV Group O risk area. ________ PRODUCT a) Red Blood Cells; b) Fresh Frozen Plasma. Recall #B-813/814-9. CODE Unit #5064631. MANUFACTURER BloodCare, doing business as Carter BloodCare, Dallas, Texas. RECALLED BY Manufacturer, by fax on November 3, 1998, and by letter dated November 16, 1998. Firm-initiated recall complete. DISTRIBUTION Texas. QUANTITY 1 unit of each component was distributed. REASON Blood products were collected from a donor who reported exposure to the Human Immunodeficiency Virus (HIV) and a diagnosis of gonorrhea. ________ PRODUCT Platelets, Pheresis. Recall #B-824-9. CODE Unit numbers: 020GE01167A, 020GE01167B, 020GE01170. MANUFACTURER American Red Cross, Missoula, Montana. RECALLED BY American Red Cross, Boise, Idaho, by telephone on or about June 15, 1998. Firm-initiated recall complete. DISTRIBUTION Montana. QUANTITY 3 units were distributed. REASON Platelets were not stored under continuous agitation, and at unmonitored temperatures.

RECALLS AND FIELD CORRECTIONS: BIOLOGICS -- CLASS III

________ PRODUCT Red Blood Cells. Recall #B-795-9. CODE Unit numbers: 26KM06849, 26KM06829, 26KG16035, 26KH42282, 26KM07605, 26FQ10019, 26KM06833, 26FM41708, 26FM41662, 26FS12649, 26FF19780, 26KM08572, 26KM08574, 26FN05156, 26FN05157. MANUFACTURER The American National Red Cross, Mobile, Alabama. RECALLED BY Manufacturer, by letter dated January 28, 1998. Firm-initiated recall complete. DISTRIBUTION Mississippi, Alabama, Florida. QUANTITY 15 units were distributed. REASON Blood products were labeled with extended expiration dates. ________ PRODUCT Fresh Frozen Plasma. Recall #B-796-9. CODE Unit numbers: 41GG09184, 41GG09187, 41GG09193, 41GG09195, 41LW67431, 41LW67476, 41FL73908, 41FN37849, 41FQ52478, 41FR38544, 41FY88316, 41GE29173, 41FE12427, 41FE12444, 41FE12445, 41FE12461, 41FE12498, 41FE12504. MANUFACTURER The American National Red Cross, Mobile, Alabama. RECALLED BY Manufacturer, by telephone and letter on April 7, 1998. Firm- initiated recall complete. DISTRIBUTION Florida. QUANTITY 18 units were distributed. REASON Blood products for which temperature after shipment was not recorded upon receipt was distributed. ________ PRODUCT Red Blood Cells. Recall #B-797-9. CODE Unit #21GL27310. MANUFACTURER The American National Red Cross, Mobile, Alabama. RECALLED BY Manufacturer, by letter dated February 16, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product for which temperature after shipment was not recorded upon receipt was distributed. ________ PRODUCT Fresh Frozen Plasma. Recall #B-802-9. CODE Unit numbers: 11FF44971, 11FM29589, 11FN66110, 11FN68771, 11FN68787, 11G27530, 11G27535, 11G28487, 11L82736, 11L84277, 11L84283, 11L84333, 11M79345, 11M79366, 11Q63435, 11Q65065, 11R36716, 11V18277, 11V18854, 11W42422. MANUFACTURER The American National Red Cross, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on February 11, 1998, followed by letter dated February 16, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 20 units were distributed. REASON Blood products for which temperature after shipment was not recorded upon receipt were distributed. ________ PRODUCT a) Whole Blood; b) Red Blood Cells. Recall #B-807/808-9. CODE Unit numbers: a) 0499840; b) 0499861. MANUFACTURER Coffee Memorial Blood Center, Inc., Amarillo, Texas. RECALLED BY Manufacturer, by letter dated March 2, 1999. Firm-initiated recall complete. DISTRIBUTION Texas. QUANTITY 1 unit of each component was distributed. REASON Blood products may have been stored at unacceptable temperatures. ________ PRODUCT Red Blood Cells. Recall #B-815-9. CODE Unit #26FJ22302. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on April 8, 1998, followed by letter dated April 13, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product was labeled with an extended expiration date. ________ PRODUCT Red Blood Cells, Irradiated. Recall #B-817-9. CODE Unit numbers: 26FJ22855, 26FJ22841, 26FF22763, 26FJ22866. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on April 16, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 4 units were distributed. REASON Blood products were labeled with extended expiration dates. ________ PRODUCT Red Blood Cells, Irradiated. Recall #B-819-9. CODE Unit #26KM11000. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on October 12, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product was labeled with an extended expiration date. ________ PRODUCT Red Blood Cells. Recall #B-820-9. CODE Unit #26KM03068. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on September 27, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product was shipped at an unacceptable temperature. ________ PRODUCT Fresh Frozen Plasma. Recall #B-821-9. CODE Unit numbers: 26FS14337, 26FM51203, 26FM51207, 26FP44904. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on August 20, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 4 units were distributed. REASON Blood products were stored at an unacceptable temperature. ________ PRODUCT Red Blood Cells, Irradiated. Recall #B-822-9. CODE Unit numbers: 26FF23540 and 26FL07437. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by telephone on June 8 and 9, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 2 units were distributed. REASON Blood products were labeled with extended expiration dates. ________ PRODUCT Red Blood Cells, Irradiated. Recall #B-823-9. CODE Unit #26KM09880. MANUFACTURER American Red Cross Blood Services, Mobile, Alabama. RECALLED BY Manufacturer, by letter dated March 18, 1998. Firm-initiated recall complete. DISTRIBUTION Alabama. QUANTITY 1 unit was distributed. REASON Blood product was labeled with an extended expiration date. ________ PRODUCT Recovered Plasma. Recall #B-825-9. CODE Unit #16LY52966. MANUFACTURER American Red Cross Blood Services, Columbus, Ohio. RECALLED BY Manufacturer, by fax on December 4, 1998. Firm-initiated recall complete. DISTRIBUTION New York. QUANTITY 1 unit was distributed. REASON Blood product corresponded to a unit of Platelets which was positive for Enterobacter aerogenes and implicated in a transfusion reaction.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

________ PRODUCT Scimed Ranger 20 mm Percutaneous Transluminal Coronary angioplasty (PTCA) Dilatation Balloon Catheter. Recall #Z-923-9. CODE Model: H74914361251, Lot numbers: 2293535 and 2293538. MANUFACTURER Boston Scientific Corporation, Ireland, Ltd., Galway, Ireland. RECALLED BY Boston Scientific /Scimed, Maple Grove, Minnesota, on April 30, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 54 catheters were distributed. REASON The wrong/unvalidated sterilization process was used for the two recalled lots. ________ PRODUCT Innovative Implants and Cover Screws. Recall #Z-926-9. CODE Catalog #OSS610, Lot #77963. MANUFACTURER Implant Innovations, Inc., Palm Beach Gardens, Florida. RECALLED BY Manufacturer, by telephone on April 30, 1999, and by letter faxed on May 11, 1999. Firm-initiated recall ongoing. DISTRIBUTION New York, Illinois, Germany, Japan. QUANTITY 25 implants. REASON Product was mislabeled as 6.0mm when it was actually 5.0mm. ________ PRODUCT Model Visulas Laser System, used in ophthalmology: a) Model No. 532; b) Model No. 532 Combi; c) Model No. 690s; d) Model No. 732S; e) Model Yag II Plus. Recall #Z-929/933-9. CODE a) Model No. 532; b) Model No. 532 Combi; c) Model No. 690s; d) Model No. 732S; e) Model Yag II Plus. MANUFACTURER Carl Zeiss, Inc., Thornwood, New York. RECALLED BY Manufacturer. FDA approved the firm's corrective action plan on May 19, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide. QUANTITY a) 20; b) 15; c) 15; d) 1; e) 35 units were distributed. REASON Device failed to comply with 21 CFR 1040.11(a)(2) in that the operator's manual lacked calibration procedures. ________ PRODUCT Model CVX-300 Excimer Laser, used in cardiology. Recall #Z-934-9. CODE Model CVX-300. (ANY SERIAL NUMBERS?) MANUFACTURER Spectranetics Corporation, Colorado Springs, Colorado. RECALLED BY Manufacturer. FDA approved the firm's corrective action plan on May 20, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide. QUANTITY 128 units were distributed. REASON The device failed to comply with 21 CFR 1040.10(f)(1) in that the protective housing had a narrow gap allowing direct line-of-sight access to the Class IV beam. ________ PRODUCT Model E2000 Ruby Laser System. Recall #Z-935-9. CODE Model E2000. MANUFACTURER Palomar Medical Products, Inc., Lexington, Massachusetts. RECALLED BY Manufacturer. FDA approved the firm's corrective action plan on May 20, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide. QUANTITY 7 units were distributed. REASON The device failed to comply with 21 CFR 1040.11(a)(2) in that the operator's manual lacked calibration procedures. ________ PRODUCT Eagle 3000 Floorloader Steam Sterilizer. Recall #Z-936-9. CODE Various serial numbers. MANUFACTURER Steris Corporation, Erie Pennsylvania. RECALLED BY Steris Corporation, Mentor, Ohio, by Customer Notifications on March 10, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide, Canada, Korea, Hong Kong, Taiwan. QUANTITY 67 units were distributed. REASON There is an error in the control system software, which causes the chamber temperatures to exceed the upper limit. ________ PRODUCT Tissue Suture Anchors, used to reattach soft tissue to bone during surgery: a) Part No. PAE5K - OBL RC5 R.0 mm Suture Anchor; b) Part No. PAS2K - PeBA S 2.8 mm Suture Anchor. Recall #Z-937/938-9. CODE Lot numbers: a) A1305B, A1205B, A1405B, A1505B, A0906B, A1006B, A1106B, A0708B, A0807BM, A1007B, A1507B, A2907B, A0211B, A0311B, A1612B, A1712B, A2701C, A2801C, A2901C, A3101C; b) A0406B, A1505B, A1905B, A0206B, A0806B, A1106B, A1806B, A1906B, A0607B, A0907B, A1607B, A0608B. MANUFACTURER Southmedic, Inc., Ontario, Canada (assembler/packager). RECALLED BY Orthopaedic Biosystems, Ltd., Inc., (OBL), Scottsdale, Arizona (specification developer/responsible firm), by letters dated January 27, 1999, February 12 and 17, 1999. Firm-initiated recall ongoing. DISTRIBUTUTION Nationwide. QUANTITY 15,162 units were distributed. REASON The sterility of the lots cannot be assured due to a packaging defect that occurred during the heat sealing process. ________ PRODUCT Portable Oxygen Generator: a) Model 5LX02; b) Model 6LX02. Recall #Z-939/940-9. CODE All codes manufactured between August 1998 through March 31, 1999. MANUFACTURER Invacare Corporation, Sanford, Florida. RECALLED BY Invacare Corporation, Elyria, Ohio., by letter on May 21, 1999. Firm-initiated field correction ongoing. DISTRIBUTION Nationwide and international. QUANTITY 29,106 units were distributed. REASON Oxygen concentrators had a high incidence of reported circuit board failures. ________ PRODUCT Patient Scale, Model RSC600. Recall #Z-941-9. CODE All units manufactured between October 1, 1998 and March 31, 1999. MANUFACTURER Invacare Corporation, Elyria, Ohio. RECALLED BY Manufacturer, by letter dated April 30, 1999. Firm-initiated recall ongoing. DISTRIBUTION Nationwide. QUANTITY 188 units were distributed. REASON Screw, which secures scale, may become disengaged, causing patient/occupant to fall from scale.

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III

________ PRODUCT a) Cobe Renalyte Rx Acid Concentrate For Bicarbonate Dialysate, 2-K and 2.5 Ca, Product Number 012-116-004; b) Cobe Renalyte Rx Acid Concentrate for Bicarbonate Dialysate, 2-K, product number 012-111-004. Recall #Z-927/928-9. CODE Lot G8H074. MANUFACTURER DiChem Concentrates, Inc., Lewisberry, Pennsylvania. RECALLED BY Gambro Healthcare, Deland, Florida, by telephone on November 5, 1998, followed by letter dated November 17, 1998. Firm-initiated recall complete. DISTRIBUTION Massachusetts, Illinois, New Jersey, West Virginia, Pennsylvania, Connecticut, New York, South Carolina, Ohio, Delaware, Missouri, Virgin Islands. QUANTITY 1,095 cases (4 one-gallon bottles per case) were distributed. REASON Some product was mislabeled as product code 012-111-004 bicarbonate dialysate. ________ UPDATE Coxial Biopsy System, manufactured by Allegiance Healthcare Corporation, Clearwater, Florida, which appeared in the May 19, 1999 Enforcement Report is a Class III recall.

END OF ENFORCEMENT REPORT FOR JUNE 2, 1999.