The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

May 10, 2000
00-19

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS I

________

PRODUCT

Chef's brand Pickled Egg-Plant, in 16-ounce jars.
Recall #F-420-0.

CODE

None.

MANUFACTURER

Bright Time Intertrade L.P., Bangkok, Thailand.

RECALLED BY

Vasinee Food Corporation, Brooklyn, New York, by letter and allergy alert on December 23, 1999. Completed recall resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets.

DISTRIBUTION

New York.

QUANTITY

200 cases (24 jars per case) were distributed.

REASON

Pickled Egg-Plant contained undeclared sulfites and sodium benzoate.

_________

PRODUCT

Babcock Hall brand Ice Cream:

a) Brownies and Raspberry Swirl Ice Cream, in bulk 3-gallon containers, 1/2 gallon, quarts, and pints
b) Vanilla Chocolate Cake Ice Cream, in bulk 3-gallon containers
c) Pumpkin Cream Pie Ice Cream, in bulk 3-gallon containers, and 1/2 gallon containers
d) Chocolate Marshmallow Ice Cream, in bulk 3-gallon containers
e) Grasshopper Ice Cream, in bulk 3-gallon containers.
Recall #F-423/427-0.

CODE

All product on the market at time recall initiation.

MANUFACTURER

Babcock Hall Dairy Plant, University of Wisconsin-Madison Department of Food Science, Madison, Wisconsin.

RECALLED BY

Manufacturer, by letter on March 16, 2000, followed by visit to relabel product on March 17, 2000. Firm-initiated field correction (relabeling) complete.

DISTRIBUTION

Wisconsin.

QUANTITY

Undetermined.

REASON

a-d) Products contain undeclared egg ingredients; e) Product contains an undeclared egg ingredient and FD&C Yellow

No. 5.

________

PRODUCT

Wing Kee Lung brand Dried Lotus Root, packaged in 6-ounce flexible plastic bags. Recall #F-419-0.

CODE

None.

MANUFACTURER

Xinhu Cereals & Oils & Export Company of Guang-Dong, Xinhui, Guangdong, China.

RECALLED BY

Tai Wing Hong Importer, Inc., Brooklyn, New York, by allergy alert and letter November 30, 1999. Completed recall resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets.

DISTRIBUTION

New York, Maryland, Pennsylvania others undetermined.

QUANTITY

50 cases (100 packages per case) were distributed.

REASON

Product contained undeclared sulfites.

________

PRODUCT

Stop & Shop brand Plain Pound Creme Cake, 15 ounces,

UPC Code 0200-05254. Recall #F-435-0.

CODE

All cakes sold on or before April 7, 2000.

MANUFACTURER

Pennant Foods, Lisle, Illinois.

RECALLED BY

The Stop & Shop Supermarket Company, Boston, Massachusetts, by E-mail and press release on April 7, 2000. Firm-initiated recall complete.

DISTRIBUTION

Connecticut, Massachusetts, New York, Rhode Island.

QUANTITY

Approximately 500 units.

REASON

Product contained undeclared eggs.

________

PRODUCT

Betty Crocker Super Moist Cherry Chip Cerisette Cake Mix, in 510g boxes. Recall #F-438-0.

CODE

F923T, F924T, K015T, K016T.

MANUFACTURER

General Mills, Inc., Toledo, Ohio.

RECALLED BY

General Mills, Inc., Minneapolis, Minnesota, by letter on March 23, 2000. The Canadian government issued a press release on March 23, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

Canada.

QUANTITY

12,136 cases (12 boxes per case) were distributed.

REASON

Product contains undeclared egg whites.

 

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS II

________

PRODUCT

Tortilla Chips and Cheese Dip Snack Kits containing a 1-7/8 ounce package of tortilla chips and a 4-1/2 ounce container of cheese dip. Recall #F-306-0.

CODE

No lot coding. Pull Date: Sell by - 30 May 00

Plant Code: 322205420.

MANUFACTURER

Frito-Lay, Inc., Dallas, Texas.

RECALLED BY

Frito-Lay, Inc., Plano, Texas, by E-mail on March 17, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

Louisiana, Mississippi, Alabama, Georgia, Florida, North Carolina, South Carolina, Tennessee, Arkansas, Kentucky, Illinois, Michigan, Indiana, Missouri, Minnesota, Wisconsin, Iowa.

QUANTITY

6,006 kits were distributed; firm estimated that 1,000 to 3,200 kits remained on market at time of recall initiation.

REASON

Product contains undeclared dairy ingredients, FD&C Yellow No. 5 and FD&C Yellow No. 6.

________

PRODUCT

Penne Pasta Salad: a) 5-ounce plastic cup; b) 4-pound food service plastic tub (brand name on 5-ounce plastic cup only: WaWa Express). Recall #F-421-0.

CODE

a) 0362 EXP 1/12/00, 0365 EXP 1/15/00;

b) 0354 EXP 1/11/00, 0362 EXP 1/19/00.

MANUFACTURER

Missa Bay, Inc., Swedesboro, New Jersey.

RECALLED BY

Manufacturer, by telephone on or before January 7, 2000, and by visit on or before January 19, 2000. Firm-initiated recall complete.

DISTRIBUTION

New Jersey, Pennsylvania, Delaware, Maryland, Virginia.

QUANTITY

a) 1,844 cups; b) 210 tubs were distributed.

REASON

Product contained undeclared sulfites.

________

PRODUCT

Babcock Hall brand Ice Cream and Sherbet:

a) Mint Ice Cream, in bulk 3-gallon, 1/2 gallon, and quarts
b) Mint Chocolate Chip Ice Cream, in bulk 3-gallon containers, 1/2 gallons, and quarts
c) Butter Pecan Ice Cream, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints
d) Chocolate Chip Cookie Dough Ice Cream, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints
>e) Banana Fudge Marble Ice Cream, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints
f) Lemon Sherbet, in bulk 3-gallon containers, 1/2 gallons, and quarts
g) Lime Sherbet, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints. Recall #F-428/435-0.

CODE

All product on the market at time recall initiation.

MANUFACTURER

Babcock Hall Dairy Plant, University of Wisconsin-Madison Department of Food Science, Madison, Wisconsin.

RECALLED BY

Manufacturer, by letter on March 16, 2000, followed by visit to relabel product on March 17, 2000. Firm-initiated field correction complete.

DISTRIBUTION

Wisconsin.

QUANTITY

Undetermined.

REASON

Products contained undeclared FD&C Yellow No. 5.

________

PRODUCT

Yellowfin Tuna Loins, frozen (row) vacuumed packaged in various weights. Recall #F-437-0.

CODE

Lot #23300; Production Dates: 12/04/98; 02/03/99; 6/08/99; 7/24/99; 8/14/99; 9/03/99; 9/06/99; 9/09/99; and 9/10/99. (Note: there may be other production dates that were already sold to customers).

MANUFACTURER/PROCESSOR

Sea Pack, Piti, Guam.

RECALLED BY

Garden and Valley Isle Seafood, Inc., Honolulu, Hawaii (distributor), by memorandum February 23 and 24, 2000, and by telephone. Firm-initiated recall ongoing.

DISTRIBUTION

Hawaii.

QUANTITY

2,262.20 pounds were distributed.

REASON

Product contained high levels of histamine and was associated with histamine poisoning.

 

RECALLS AND FIELD CORRECTIONS: FOODS -- CLASS III

________

PRODUCT

Sangria Senorial Natural and Artificial Flavor Sangria Flavored Carbonated Beverage, in 11.16 fluid ounce (330mL) bottles. Recall #F-422-0.

CODE

J13B92 02:47.

MANUFACTURER

Novamex de Mexico, Cd. Juarez Chih., Mexico.

RECALLED BY

Tipp Enterprises, doing business as Novamex, El Paso, Texas, by telephone and fax on or about April 30, 1999. Firm-initiated recall complete.

DISTRIBUTION

Nationwide.

QUANTITY

100,000 cases (24 per case) were distributed.

REASON

Product failed to bear an ingredient statement.

________

PRODUCT

Tuv Taam brand Cole Slaw, in 14-ounce rigid plastic containers. Recall #F-436-0.

CODE

All codes.

MANUFACTURER

Tuv Taam Corporation, Brooklyn, New York.

RECALLED BY

Manufacturer, by visit starting on December 8, 1999. Completed recall resulted from sample analysis and follow by the New York State Department of Agriculture and Markets.

DISTRIBUTION

New York and New Jersey.

QUANTITY

Approximately 15,000 packages were distributed.

REASON

Product contained undeclared potassium sorbate.

 

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS I

________

PRODUCT

Cypress Medical Products Sterile Procedure Kits and Tray containing antiseptic products recalled by Clinipad Corporation. Recall #Z-522-0.

CODE

a) Kits containing Clinidine Ointment, Sterile, 1.0 gram

I.V. Start Kits:

Reorder #82-03, Economy, 50 kits/case, lot 80410, and all lots starting with "9" or "0"

Reorder #82-05, Deluxe, 50 kits/case, all lots starting with "9" or "0"

Reorder #82-06, Latex-free, 50 kits/case, all lots starting with "9" or "0"

Reorder #82-07, Premium, 50 kits/case, lots 71103, 71104, 80409, all lots starting with "9" or "0".

Daily Catheter Care Trays, 20 trays/case, reorder #86-101, lot 71205, 71208, 90403, 90225.

b) Kits containing Clinidine Ointment - 4339SC, Clinguard

Protective Dressing Prep Pad - 8133S-B, Cliniswab

Isopropyl Alcohol Antiseptic Swabsticks - 1245SB, and Clinipad Alcohol Prep - 0110:

Omnicare Central Line Dressing Kit, reorder #84-117, lots 71215 and 80404;

c) Kits containing Clinipad Alcohol Prep - 0110:

Sharp Debridement Package, 20 kits/case, order number SDP, reorder #84-160, lots 70812 and 70813;

d) Kits containing Clinipad Iodophor PVP Iodine USP Prep Pads - 9153:

Skin Staple Remover kits:

Reorder #82-15, Premium, 50 kits/case, lot 71016

Reorder #1295-815, Starline, 50 kits/case, lot 70903

Reorder #241, PSS Select, 50 kits/case, lot 80214

Suture Removal Kit, reorder #81-213, lot 80305.

MANUFACTURER

Cypress Medical Products, Ltd., McHenry, Illinois.

RECALLED BY

Manufacturer, by letter on March 22, 2000. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide.

QUANTITY

96,736 kits/trays were distributed.

REASON

The antiseptic products may be contaminated with bacteria.

________

PRODUCT

Various Sterile Procedure Kits containing antiseptic products recalled by Clinipad Corporation:

ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with

T-Fasteners, Complete Kit, 14 French (List #00227);

ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with

T-Fasteners, Complete Kit, 18 French (List #00238);

ROSS Flexiflo Introducer Gastrostomy Kit with Brown/Mueller

T-Fastener Set, Complete Kit, 18 French (List #50190);

Ross Inverta-PEG Pull Technique Gastrostomy Kit with Roll-Tip Bumper, Complete Kit with Snare, 20 French (List #52002);

Ross Inverta-PEG Over-the-Guidewire Gastrostomy Kit with Roll-Tip Bumper, Complete Kit with Snare, 20 French (List #52354);

ROSS Flexiflow Versa-PEG Gastrostomy Kit, Complete Kit, 22 French (list #50520);

ROSS Flexiflo Inverta-PEG Over-the-Guidewire Gastrostomy Kit with Roll-Tip Bumper, Complete Kit with Snare, 20 French (List #51492) [Obsolete]. Recall #Z-523-0.

CODE

All lot numbers beginning with 25 through 62.

MANUFACTURER

Ross Products Division, Abbott Laboratories, Inc., Columbus, Ohio.

RECALLED BY

Manufacturer, by letters dated March 20 and 27, 2000, and providing stickers to be placed on kits advising users not to use the Clinipad products and to destroy them. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide, Chile, Hungary, Korea, Mexico, Netherlands, Singapore, Turkey.

QUANTITY

77,492 (including the Class II portion).

REASON

Kits contain Clinipad antiseptic products which may be contaminated with bacteria.

 

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS II

________

PRODUCT

Sterile Procedure Kits containing antiseptic products recalled by Clinipad Corporation:

ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with Reinforced Webbed Bumper, Complete Kit, 16 French (List #50724); ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with Reinforced Webbed Bumper, Complete Kit, 20 French (List #50738). Recall #Z-523-0.

CODE

All lot numbers beginning with 25 through 62.

MANUFACTURER

Ross Products Division, Abbott Laboratories, Inc., Columbus, Ohio.

RECALLED BY

Manufacturer, by letters dated March 20 and 27, 2000, and providing stickers to be placed on kits advising the users not to use the Clinipad products and to destroy them. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide, Chile, Hungary, Korea, Mexico, Netherlands, Singapore, Turkey.

QUANTITY

77,492 (including the Class I portion).

REASON

Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

Heart Valve Allograft Models AVOO, PVOO, AVO5.

Recall #Z-628/631-0.

CODE

Serial Numbers: 3887502, 3887391, 6100597, 6276377.

MANUFACTURER

Cryolife, Inc., Kennesaw, Georgia.

RECALLED BY

Manufacturer, by letter on March 23 and 27, 2000. Firm-initiated recall complete.

DISTRIBUTION

Tennessee, Oklahoma, Michigan.

QUANTITY

4 valves.

REASON

Donor did not meet current guidelines regarding serodilution of plasma.

________

PRODUCT

Hemodialysis/Apheresis Trays containing recalled Clinipad products.

Recall #Z-632-0.

CODE

All items listed below are Vas-Cath products.

5500102

Soft-Cell Catheterless Tray

5500103

Soft-Cell Tray 12.5F x 19cm

5500114

Flexxicon II Catheterless Tray

5500125

Soft-Cell Tray 12.5F x 27cm

550012

Soft-Cell Tray 12.5F x 15cm

5500130

Soft-Cell Tray 12.5F x 17cm

5500132

Soft-Cell Tray 12.5F x 27cm

5525150

Flexxicon II Tray 12F x 15cm

5525200

Flexxicon II Tray 12F x 20cm

5525240

Flexxicon II Tray 12F x 24cm

5526120

Flexxicon II 12.5 cm Tray

5526150

Flexxicon II Tray 12F x 15cm

5526200

Flexxicon II Tray 12F x 20cm

5526620

Flexxicon II Tray 12F x 12.5cm

5526650

Flexxicon II Tray 12F x 15cm

5526700

Flexxicon II Tray 12F x 20cm

5535120

Soft-Cell Tray 12.5 x 12cm

5535190

Soft-Cell Tray 12.5F x 19cm

5535230

Soft-Cell Tray 12.5F x 23cm

5535620

Soft-Cell Tray 12.5F x 12cm

5535690

Soft-Cell Tray 12.5F x 19cm

5535730

Soft-Cell Tray 12.5F x 23cm

5536120

Soft-Cell Tray 12.5F x 12cm

5536190

Soft-Cell Tray 12.5F x 19cm

5536230

Soft Cell Pre-Curved Dual Lumen Chronic Catheter

5545150

Vaccess Tray 8F x 15cm

5545200

Vaccess Tray 8F x 20cm

5545700

Vaccess Tray 8F x 20cm

5591000

Niagara Catheterless Tray

5595150

Niagara Tray 13.5F x 15cm

5595200

Niagara Tray 13.5F x 20cm

5595240

Niagara Tray 13.5F x 24cm

5596120

Niagara Tray 13.5F x 12.5cm

5596150

Niagara Tray 13.5F x 15cm

5596200

Niagara Tray 13.5F x 20cm

6531000

Opti-Flow Catheterless Tray

5635192

Opti-Flow Tray 14.5F x 19cm

5635692

Opti-Flow Tray 14.5F x 19cm

5635732

Opti-Flow Tray 14.5F x 23cm

5635232

Opti-Flow Tray 14.5F x 23cm

5636192

Opti-Flow Tray 14.5F x 19cm

5636232

Opti-Flow Tray 14.5F x 23cm.

MANUFACTURER

Bard Access Systems, Inc., Salt Lake City, Utah.

RECALLED BY

Manufacturer, by letters on March 20 and 22, 2000. Customers were given the option of returning any affected product for credit or removing the Clinipad components. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad, Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide.

QUANTITY

108,347 kits/trays were distributed.

REASON

Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

LSL Sterile Mid-Stream Urine Specimen Kit, 120 ml, containing recalled Clinipad product. Recall #Z-633-0.

CODE

Kit Reorder #LSL1030, All lots beginning with 7, 8, 9, and 0A3006, 0B3007.

MANUFACTURER

LSL Industries, Inc., Chicago, Illinois.

RECALLED BY

Manufacturer, by letters dated March 20, 1000, and April 6, 2000, instructing users to hold kits in quarantine until stickers were provided to be placed on the kits advising the user not to use the Clinipad product and to destroy it. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Tennessee, Kansas, Vermont, Texas, Michigan, Pennsylvania, New York, Oklahoma, Massachusetts, Indiana, Florida, New Jersey, Oregon, Arizona, Alaska.

QUANTITY

64,400 kits were distributed.

REASON

Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

Deltec Procedural Tray (containing recalled Clinipad products) for use with CliniCath Peripherally Inserted Catheters, Product Numbers 21-2501-01 and 21-2515-01; Deltec CliniCath Peripherally Inserted Catheters, with Tray, Product Numbers 21-2508-01, 21-2509-01, 21-2510-01, 21-2511-01, 21-2512-01, and 21-2513-01. Recall #Z-634-0.

CODE

All lots distributed since 1/1/97.

MANUFACTURER

Sims Deltec, Inc., St. Paul, Minnesota.

RECALLED BY

Manufacturer, by letter dated April 5, 2000. Firm stated that consignees could keep CliniCath products needed for urgent patient needs and return the rest. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide and international.

QUANTITY

3,512 trays were distributed.

REASON

Trays contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

Zipprep Electrode Starter Kits containing Clinipad Alcohol Antiseptic Swabsticks, intended for use with EEG monitoring Systems. Recall #Z-635-0.

CODE

Aspect Product #185-0043; Aspect Lot Numbers: 418442, 413659, 423659, 417506.

MANUFACTURER

Aspect Medical Systems, Inc., Natick, Massachusetts.

RECALLED BY

Manufacturer, by notifying customers on March 29, 2000 to discontinues and destroy any products on hand. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Washington state, Florida, West Virginia, California, Ohio, Massachusetts, Pennsylvania, Louisiana, Kentucky, and international.

QUANTITY

81 kits (all past expiration dating).

REASON

Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

General Procedure Kit. Recall #Z-637-0.

CODE

Catalog DYNJTS0300, Lot 9904007

DYNJTS4302, Lot 9904002

DYNJTS0400, Lot 9903041

DYNJTS4306, Lot 9903037

DYNJTS7025, Lot 9903035

DYNJTS4301, Lot 9904001

DYNJTS4303, Lot 9903036.

MANUFACTURER

Physician Industries Manufacturing, West Valley City, Utah.

RECALLED BY

Manufacturer, by telephone on May 6, 1999. Firm-initiated recall complete.

DISTRIBUTION

Georgia, Illinois, Nevada, Pennsylvania, Texas.

QUANTITY

2,880 kits were distributed.

REASON

Distributed kits had positive bioburden test results after EtO sterilization.

________

PRODUCT

Allograft Heart Valve, Model AVOO and PVOO.

Recall #Z-638/639-0.

CODE

Serial Numbers: 3887693, 3887707, 3895343, 3895351, 6004946, 6004954, 6262155, 6268331, 6297846.

MANUFACTURER

Cryolife, Inc., Kennesaw, Georgia.

RECALLED BY

Manufacturer, by letters on April 3 and 6, 2000. Firm-initiated recall complete.

DISTRIBUTION

New York, Colorado, Tennessee, Florida, Michigan, Maryland, Ohio, California.

QUANTITY

9 valves.

REASON

Donor did not meet current guideline regarding serodilution of plasma.

________

PRODUCT

Hermitage Eye Pads, Sterile 2 1/8" X 2 5/8", Product of China. Sterilized in the U.S.A, for use as a bandage over the eye for protection or absorption of secretions: a) Item No. EP-1600; b) Item No. EP-2000. Recall #Z-640/641-0.

CODE

Lot #5588.

MANUFACTURER

Shaoxing Life Surgical Dressing Company, Ltd., Zhejiang Prov. Shaoxing County, China.

RECALLED BY

Hermitage Hospital Products, Inc., Niantic, Connecticut, by telephone on April 19, 2000, followed by fax. Firm-initiated recall ongoing.

DISTRIBUTION

Arkansas, California, Oklahoma, Puerto Rico.

QUANTITY

a) 33 cases (600 per case); b) 33 cases (2000 per case).

REASON

Product was not sterilized and incorrectly transferred to released inventory without being gamma sterilized.

________

PRODUCT

Allograft Heart Valve, Models PV00, AV00, AV40, MV10 PV05.

Recall #Z-642/646-0.

CODE

Serial Numbers: 3889149, 3911446, 3911454, 6965792, 6965768, 6313744.

MANUFACTURER

Cryolife, Inc., Kennesaw, Georgia.

RECALLED BY

Manufacturer, by letters on April 7, 12, and 13, 2000. Firm-initiated recall complete.

DISTRIBUTION

New York, Pennsylvania, Louisiana, California, North Carolina, Canada.

QUANTITY

6 valves.

REASON

Donors did not meet current guidelines regarding serodilution of plasma.

________

PRODUCT

Disposable Coronary Control Syringe, 20mL,

Catalog #K01-04072. Recall #Z-647-0.

CODE

Lot #A140030.

MANUFACTURER

Merit Medical Systems, Inc., South Jordan, Utah.

RECALLED BY

Manufacturer, by telephone on January 21, 2000. Firm-initiated recall complete.

DISTRIBUTION

South Carolina.

QUANTITY

25 syringes.

REASON

Syringes were shipped to customer prior to terminal sterilization processing.

________

PRODUCT

Flo-Rester Sizer Pack (of blood vessel sizers), used to determine the internal diameters of arteries and veins:

Catalog Numbers: a) FRSZPK05; b) FRSZ400; c) FRSZ300;

d) FRSZ250; e) FRSZ200; f) FRSZ150. Recall #Z-648/653-0.

CODE

Lot Numbers: a) 526069, 525165, 523878, 522704, 521864, 518549, and 515997; b) 526067, 525588, 524943, 520805, and 516003;

c) lot number 516002; d) 516001; e) 516000; f) 515999.

MANUFACTURER

Bio-Vascular, Inc., St. Paul, Minnesota.

RECALLED BY

Manufacturer, by letter dated April 19, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

Nationwide, Brazil, Canada, Denmark, Germany, United Kingdom.

QUANTITY

107 sizers were distributed.

REASON

The bulb of the device may loosen and become detached from the shaft.

________

PRODUCT

VACUTAINER Urine Collection Kit for Midstream Specimens containing recalled Clinipad products.

Recall #Z-659-0.

CODE

Catalog Number 364952 Vacutainer Urine Collection kit with

sterile cup, 5ml preservative tube, and Clinipad Rantex

Wipe, Lot 9E165, 9F229, 9H142, 9J205, 9J206;

Catalog Number 364972 Vacutainer Urine Collection kit with

sterile cup, 12 ml conical urinalysis tube, 5 ml

preservative tube, and Clinical Rantex Wipe, Lot 9E170,

9E171, 9F233, 9G085, 9H147, 9J201.

MANUFACTURER

Becton Dickinson & Company, Broken Bow, Nebraska.

RECALLED BY

Becton Dickinson & Company, Franklin Lakes, New Jersey, by letter dated April 7, 2000, instructing customers to place warning stickers on kits. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide.

QUANTITY

1,428,050 kits were distributed.

REASON

Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

FLOW-20, used as a gastrointestinal feeding tube:

a) Flow-20-Push; b) Flow-20-Pull; c) Flow-20-Pull-S.

Recall #Z-660/662-0.

CODE

Lot Numbers: a) 1276176 and 1280434; b) 1276560;

c) 1277439 and 1278396.

MANUFACTURER

Wilson-Cook Medical, Inc., Winston-Salem, North Carolina.

RECALLED BY

Manufacturer, by telephone, fax and mail beginning November 10, 1999. Firm-initiated recall complete.

DISTRIBUTION

Alabama, Arizona, California, Connecticut, Florida, Louisiana, Massachusetts, Missouri, New Hampshire, New York, North Dakota, Ohio, Pennsylvania, Texas, Vermont, Wisconsin, Australia, Hungary, Jordan, Singapore.

QUANTITY

246 units were distributed.

REASON

Kits may contain a larger external bumper, which may not properly secure the feeding tube.

________

PRODUCT

Magnetom 42 SP Magnetic Resonance Imaging System with

Software Version Number 2A2.7B. Computer aided medical device generates magnetic field resulting in a digital image of the body:

a) Magnetom 42 SP Magnetic Resonance Imaging Systems, Model No. 89 69 131 (SP-A);

b) Magnetom 42 SP Magnetic Resonance Imaging Systems, Model No. 87-29-485 (SP). Recall #Z-664/665-0.

CODE

Each unit is identified with a unique serial number. Those units sold (194 units) in the USA with the software version 2A2.7B are listed as follows:

03010

03011

03014

03015

03020

03022

03023

03026

03027

03034

03036

03047

03048

03054

03055

03058

03059

03060

03062

03063

03064

03066

03067

03068

03069

03070

03073

03084

03088

03091

03094

03095

03096

03097

03099

03102

03103

03115

03116

03120

03121

03125

03130

03139

03141

03144

03146

03148

03155

03157

03158

03160

03164

03168

03169

03179

03185

03186

03188

03189

03202

03203

03205

03209

03212

03213

03215

03218

03222

03227

03228

03229

03230

03231

03234

03235

03236

03237

03238

03239

03240

03242

03245

03256

03257

03258

03261

03262

04004

04015

04016

04017

04019

04020

04026

04027

04028

04031

04033

04036

04037

04038

04041

04042

04044

04045

04046

04047

04049

04052

04053

04058

04059

04063

04064

04068

04070

04076

04080

04082

04083

04084

04087

04094

04095

04096

04099

04102

04105

04107

04108

04114

04115

04117

04119

04121

04126

04127

04128

04136

04137

04142

04147

04153

04154

04156

04157

04160

04162

04163

04164

04166

04167

04169

04170

04171

04172

04173

04176

04180

04189

04190

04192

04201

04204

04208

04209

04210

04211

04213

04214

04215

04218

04220

04229

04238

04244

04248

04258

04259

04265

04266

04268

04273

04274

04277

04278

04279

04280

04285

04289

04290

04292

04296

   

Those units sold overseas (193 units) are with the affected software version are listed below:

03006

03013

03018

03019

03024

03025

03031

03033

03039

03042

03045

03046

03049

03051

03057

03061

03071

03072

03078

03080

03081

03082

03083

03086

03093

03108

03109

03111

03112

03113

03114

03118

03119

03122

03124

03126

03127

03129

03131

03132

03133

03135

03143

03145

03147

03151

03153

03154

03165

03166

03170

03171

03173

03182

03183

03184

03187

03190

03191

03208

03217

03224

03226

03244

03246

03260

04002

04003

04005

04007

04008

04012

04013

04018

04021

04023

04024

04025

04029

04030

04032

04039

04040

04043

04048

04050

04055

04056

04060

04061

04062

04065

04066

04069

04071

04072

04073

04074

04075

04077

04078

04079

04086

04088

04090

04092

04093

04097

04100

04104

04109

04110

04111

04113

04118

04124

04125

04129

04130

04131

04132

04133

04134

04139

04143

04144

04146

04148

04149

04150

04152

04158

04175

04177

04178

04179

04181

04183

04184

04185

04186

04191

04193

04194

04195

04197

04198

04199

04200

04205

04206

04216

04217

04221

04223

04224

04225

04228

04233

04234

04235

04236

04246

04247

04249

04250

04251

04252

04253

04254

04255

04256

04257

04260

04261

04262

04263

04264

04269

04271

04272

04275

04276

04282

04283

04288

04291

04294

04295

04301

04302

04303

04304.

     

MANUFACTURER

Siemens AG, Erlangen, Germany.

RECALLED BY

Siemens Medical Systems, Inc., Iselin, New Jersey, by letter dated February 24, 2000. Firm-initiated field correction ongoing.

DISTRIBUTION

Nationwide and international.

QUANTITY

Approximately 387 units.

REASON

Patients who are registered on the MRI systems with birth dates of January 1, 2000 or later will be misinterpreted by the software to have been born in 1900.

________

PRODUCT

Product insert for Acetaminophen (Enzymatic Rate) Reagent, used for the quantitative determination of Acetaminophen in serum, Catalog #502-13. Recall #Z-667-0.

CODE

Catalog No. 502-13.

MANUFACTURER

Diagnostic Chemicals Ltd., Charlottestown, PE, Canada.

RECALLED BY

Manufacturer, by letters dated March 15 and 21, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

Nationwide and Canada

QUANTITY

310 units were distributed.

REASON

The "Product Insert" does not contain information on interference from N-acetylcysteine (NAC), the drug of choice for treating acetaminophen overdose.

_______

UPDATE

Abbott ARCHITECT i System with Software Version 1.0 Processing Module, List No. 08C89-00-1, designed to perform automated immunoassay tests bassed on the use of Chemiluminescent Microparticle Immunoassay (CMIA) detection technology. Recall #Z-611-0, which appeared in the May 3, 2000 Enforcement Report should read: RECALLED BY: Manufacturer, by telephone on February 25, 1999.

 

RECALLS AND FIELD CORRECTIONS: DEVICES -- CLASS III

________

PRODUCT

GAP IgG Helicobacter Pylori Enzyme Immunoassay Kit, for the qualitative detection of IgG specific antibodies to helicobacter pylori in human serum. Recall #Z-610-0.

CODE

Catalog #404 2050, Lot #0426753-1/2 EXP 11/00.

MANUFACTURER

Biomerica, Inc., Newport Beach, California.

RECALLED BY

Manufacturer, by telephone on or about March 17, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

California.

QUANTITY

252 kits were distributed.

REASON

The stability parameters were starting to draft and would be out-of-specification before the expiration date.

________

PRODUCT

Endoscopic Blunt Dissecting Instrument, used to probe or move organs during surgery. Recall #Z-636-0.

CODE

KJJU02, KJJU03, BKJU01, BKJU02.

MANUFACTURER

Manufacturas Frontera Norte SA de CV, CD Juarez, Mexico.

RECALLED BY

Eagle Endoscopic, Inc., Bartonville, Texas, by telephone on April 14, 2000, and by telephone followed by fax on April 17, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

Illinois and Tennessee.

QUANTITY

473 cases (20 units per case) were distributed.

REASON

Tip separated from body while in use.

________

PRODUCT

Merge ExamWorks Video Units, used with equipment for connecting diverse diagnostic imaging products: Model/Catalog Numbers: 006-04100-00, 006-04100-01, 006-04100-00S, 006-04100-01S, and 006-04100-02. Recall #Z-654/658-0.

CODE

All units shipped prior to March 30, 2000.

MANUFACTURER

Nicolet, Madison, Wisconsin.

RECALLED BY

Merge Technologies, Inc., West Allis, Wisconsin, by "Important Customer Product Advisory" dated April 21, 2000. Firm-initiated field correction ongoing.

DISTRIBUTION

California, Connecticut, Kansas, Minnesota, Missouri, Nebraska, Ohio, Oregon, Virginia, Wisconsin, Brazil, Netherlands, Sweden.

QUANTITY

103 units were distributed.

REASON

There is a potential for horizontal banding to appear on video screen images and images printed on laser film.

________

PRODUCT

Kodak Oncology Film, identified as ECFLM film, 35 x 43 cm, Catalog No. 8715757. Recall #Z-663-0.

CODE

Emulsion Numbers: 0116-022-12, 0116-022-14, 0116-022-16.

MANUFACTURER

Eastman Kodak Company, Windsor, Colorado.

RECALLED BY

Eastman Kodak Company, Rochester, New York, by telephone on March 24, 2000, and letters dated April 13, 2000. Firm-initiated recall ongoing.

DISTRIBUTION

California, Missouri, North Carolina, Pennsylvania, Texas, Wisconsin.

QUANTITY

26 boxes (100 sheets per box) were distributed.

REASON

Graphic Arts film was slit, chopped, and packaged as Oncology X-ray Film.

________

PRODUCT

Ampicillin/Sulbactam 10/10 mcg Antimicrobial Susceptibility Test Disk, packaged 50 disks per box, intended for use in the standardized disk susceptibility test and is the recommended method of performing antimicrobial disk diffusion testing. Recall #Z-666-0.

CODE

Product #33-135, Lot #216785 EXP 4/02.

MANUFACTURER

Oxoid Ltd., Basingstoke Hants RG248PW, United Kingdom.

RECALLED BY

Remel Inc., Lenexa, Kansas, by telephone during March 13 - 15, 2000. Firm-initiated recall complete.

DISTRIBUTION

Nationwide.

QUANTITY

100 packages were distributed.

REASON

The zones of inhibition obtained were below the acceptable zone diameter of the NCCLS Standard for Antimicrobial Disk Susceptibility Tests.

 

RECALLS AND FIELD CORRECTIONS: VETERINARY PRODUCTS -- CLASS II

________

PRODUCT

Bovine mastitis treatment kits containing Clinipad alcohol pads:

a) U.S. Vet Masti-Clear Penicillin G Procaine in Sesame Oil Lactating Cow Mastitis treatment. GC Hanford - firm on label.

b) U.S. Vet go-dry. Penicillin G Procaine in Sesame Oil Dry Cow Mastitis Treatment - GC Hanford - firm on label

c) First Choice Dry Cow Mastitis Treatment, Penicillin G Procaine in Sesame Oil - KenAG Inc., Ashland , O H- firm on label

d) Aqua-Mast II Lactating Cow Mastitis Treatment Penicillin G Procaine in Sesame Oil. KenAG, Inc., Ashland, OH - firm on label

e) Amoxi-Mist Lactating Cow Formula - Rx veterinary product - Pfizer Animal Health, Exton, PA - firm on label

f) Dariclox Lactating Cow Formula - Rx veterinary product - Pfizer Animal Health Exton, PA - firm on label

g) Orbenin Quick Release Cloxacillin Sodium Intramammary Infusion USP.Veterinary product - Canadian market - SmithKline Beecham Animal Health, Mississauga, Ontario - firm on label.

h) Orbenin-DC Dry Cow Intramammary Infusion - Rx veterinary product - Pfizer Animal Health, Exton, PA firm on label.

Recall #V-082/089-0.

CODE

All lots produced from 3/7/97 to 3/9/2000 - Kits produced AFTER 3/9/2000 will have Webcol Alcohol Preps.

MANUFACTURER

G.C. Hanford Manufacturing Company, Syracuse, New York.

RECALLED BY

Manufacturer, by letter mailed on March 29, 2000, requesting consignees to remove and destroy all Clinipad alcohol preps packaged with the kits. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide.

QUANTITY

619,820 cartons were distributed.

REASON

Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.

________

PRODUCT

a) Albadry Plus Suspension Kits - Penicillin G procaine and novobiocin sodium dry cow formula. 12-10 ml syringes and clinipad alcohol pads for intramammary infusion for treatment of mastitis;

b) Albadry Plus Suspension Kits - Penicillin G procaine and novobiocin sodium dry cow formula. 144-10 ml syringes and clinipad alcohol pads for intramammary infusion for treatment of mastitis. Recall #V-090/091-0.

CODE

All lots - Clinipad lot numbers packaged in kits - 705751, 801134, 803577, 803755.

MANUFACTURER

Pharmacia & Upjohn, Kalamazoo, Michigan.

RECALLED BY

Manufacturer, by letter mailed on April 6, 2000. Firm-initiated recall ongoing.

COMPONENT MANUFACTURER

Clinipad, Corporation, Rocky Hill, Connecticut.

DISTRIBUTION

Nationwide.

QUANTITY

176,000 cartons of 12 and 5,850 pails of 144 were distributed.

REASON

Kits contain Clinipad Alcohol Pads labeled as sterile. Clinipad is unable to assure sterility.

 

MEDICAL DEVICE SAFETY ALERTS:

________

PRODUCT

Abbott Ambulatory Infusion Manager Plus Single-Channel Infusion Pump, five types of programming options available for TPN, pain management, intermittent, variable time and contiuous delivery of medications and nutritional fluids.

Safety Alert #N-014-0.

CODE

List #13967, Serial Numbers 12300001/12320000.

MANUFACTURER

Abbott Laboratories, Inc., San Diego, California.

ALERTED BY

Abbott Laboratories, Inc., Abbott Park, Illinois, by letter date, by letter on March 2, 2000.

DISTRIBUTION

Nationwide.

QUANTITY

15,947 units were distributed.

REASON

There is a possibility that medication therapy may be missed on March 1, 2000, due to the leap year, if the pumps were not appropriately reprogrammed on March 1, 2000.

END OF ENFORCEMENT REPORT FOR MAY 10, 2000

####