May 10, 2000
00-19
________
PRODUCT
Chef's brand Pickled Egg-Plant, in 16-ounce jars.
Recall #F-420-0.
CODE
None.
MANUFACTURER
Bright Time Intertrade L.P., Bangkok, Thailand.
RECALLED BY
Vasinee Food Corporation, Brooklyn, New York, by letter and allergy alert on December 23, 1999. Completed recall resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets.
DISTRIBUTION
New York.
QUANTITY
200 cases (24 jars per case) were distributed.
REASON
Pickled Egg-Plant contained undeclared sulfites and sodium benzoate.
_________
PRODUCT
Babcock Hall brand Ice Cream:
a) Brownies and Raspberry Swirl Ice Cream, in bulk 3-gallon containers, 1/2 gallon, quarts, and pints
b) Vanilla Chocolate Cake Ice Cream, in bulk 3-gallon containers
c) Pumpkin Cream Pie Ice Cream, in bulk 3-gallon containers, and 1/2 gallon containers
d) Chocolate Marshmallow Ice Cream, in bulk 3-gallon containers
e) Grasshopper Ice Cream, in bulk 3-gallon containers.
Recall #F-423/427-0.
CODE
All product on the market at time recall initiation.
MANUFACTURER
Babcock Hall Dairy Plant, University of Wisconsin-Madison Department of Food Science, Madison, Wisconsin.
RECALLED BY
Manufacturer, by letter on March 16, 2000, followed by visit to relabel product on March 17, 2000. Firm-initiated field correction (relabeling) complete.
DISTRIBUTION
Wisconsin.
QUANTITY
Undetermined.
REASON
a-d) Products contain undeclared egg ingredients; e) Product contains an undeclared egg ingredient and FD&C Yellow
No. 5.
________
PRODUCT
Wing Kee Lung brand Dried Lotus Root, packaged in 6-ounce flexible plastic bags. Recall #F-419-0.
CODE
None.
MANUFACTURER
Xinhu Cereals & Oils & Export Company of Guang-Dong, Xinhui, Guangdong, China.
RECALLED BY
Tai Wing Hong Importer, Inc., Brooklyn, New York, by allergy alert and letter November 30, 1999. Completed recall resulted from sample analysis and follow-up by the New York State Department of Agriculture and Markets.
DISTRIBUTION
New York, Maryland, Pennsylvania others undetermined.
QUANTITY
50 cases (100 packages per case) were distributed.
REASON
Product contained undeclared sulfites.
________
PRODUCT
Stop & Shop brand Plain Pound Creme Cake, 15 ounces,
UPC Code 0200-05254. Recall #F-435-0.
CODE
All cakes sold on or before April 7, 2000.
MANUFACTURER
Pennant Foods, Lisle, Illinois.
RECALLED BY
The Stop & Shop Supermarket Company, Boston, Massachusetts, by E-mail and press release on April 7, 2000. Firm-initiated recall complete.
DISTRIBUTION
Connecticut, Massachusetts, New York, Rhode Island.
QUANTITY
Approximately 500 units.
REASON
Product contained undeclared eggs.
________
PRODUCT
Betty Crocker Super Moist Cherry Chip Cerisette Cake Mix, in 510g boxes. Recall #F-438-0.
CODE
F923T, F924T, K015T, K016T.
MANUFACTURER
General Mills, Inc., Toledo, Ohio.
RECALLED BY
General Mills, Inc., Minneapolis, Minnesota, by letter on March 23, 2000. The Canadian government issued a press release on March 23, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
Canada.
QUANTITY
12,136 cases (12 boxes per case) were distributed.
REASON
Product contains undeclared egg whites.
________
PRODUCT
Tortilla Chips and Cheese Dip Snack Kits containing a 1-7/8 ounce package of tortilla chips and a 4-1/2 ounce container of cheese dip. Recall #F-306-0.
CODE
No lot coding. Pull Date: Sell by - 30 May 00
Plant Code: 322205420.
MANUFACTURER
Frito-Lay, Inc., Dallas, Texas.
RECALLED BY
Frito-Lay, Inc., Plano, Texas, by E-mail on March 17, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
Louisiana, Mississippi, Alabama, Georgia, Florida, North Carolina, South Carolina, Tennessee, Arkansas, Kentucky, Illinois, Michigan, Indiana, Missouri, Minnesota, Wisconsin, Iowa.
QUANTITY
6,006 kits were distributed; firm estimated that 1,000 to 3,200 kits remained on market at time of recall initiation.
REASON
Product contains undeclared dairy ingredients, FD&C Yellow No. 5 and FD&C Yellow No. 6.
________
PRODUCT
Penne Pasta Salad: a) 5-ounce plastic cup; b) 4-pound food service plastic tub (brand name on 5-ounce plastic cup only: WaWa Express). Recall #F-421-0.
CODE
a) 0362 EXP 1/12/00, 0365 EXP 1/15/00;
b) 0354 EXP 1/11/00, 0362 EXP 1/19/00.
MANUFACTURER
Missa Bay, Inc., Swedesboro, New Jersey.
RECALLED BY
Manufacturer, by telephone on or before January 7, 2000, and by visit on or before January 19, 2000. Firm-initiated recall complete.
DISTRIBUTION
New Jersey, Pennsylvania, Delaware, Maryland, Virginia.
QUANTITY
a) 1,844 cups; b) 210 tubs were distributed.
REASON
Product contained undeclared sulfites.
________
PRODUCT
Babcock Hall brand Ice Cream and Sherbet:
a) Mint Ice Cream, in bulk 3-gallon, 1/2 gallon, and quarts
b) Mint Chocolate Chip Ice Cream, in bulk 3-gallon containers, 1/2 gallons, and quarts
c) Butter Pecan Ice Cream, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints
d) Chocolate Chip Cookie Dough Ice Cream, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints
>e) Banana Fudge Marble Ice Cream, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints
f) Lemon Sherbet, in bulk 3-gallon containers, 1/2 gallons, and quarts
g) Lime Sherbet, in bulk 3-gallon containers, 1/2 gallons, quarts, and pints. Recall #F-428/435-0.
CODE
All product on the market at time recall initiation.
MANUFACTURER
Babcock Hall Dairy Plant, University of Wisconsin-Madison Department of Food Science, Madison, Wisconsin.
RECALLED BY
Manufacturer, by letter on March 16, 2000, followed by visit to relabel product on March 17, 2000. Firm-initiated field correction complete.
DISTRIBUTION
Wisconsin.
QUANTITY
Undetermined.
REASON
Products contained undeclared FD&C Yellow No. 5.
________
PRODUCT
Yellowfin Tuna Loins, frozen (row) vacuumed packaged in various weights. Recall #F-437-0.
CODE
Lot #23300; Production Dates: 12/04/98; 02/03/99; 6/08/99; 7/24/99; 8/14/99; 9/03/99; 9/06/99; 9/09/99; and 9/10/99. (Note: there may be other production dates that were already sold to customers).
MANUFACTURER/PROCESSOR
Sea Pack, Piti, Guam.
RECALLED BY
Garden and Valley Isle Seafood, Inc., Honolulu, Hawaii (distributor), by memorandum February 23 and 24, 2000, and by telephone. Firm-initiated recall ongoing.
DISTRIBUTION
Hawaii.
QUANTITY
2,262.20 pounds were distributed.
REASON
Product contained high levels of histamine and was associated with histamine poisoning.
________
PRODUCT
Sangria Senorial Natural and Artificial Flavor Sangria Flavored Carbonated Beverage, in 11.16 fluid ounce (330mL) bottles. Recall #F-422-0.
CODE
J13B92 02:47.
MANUFACTURER
Novamex de Mexico, Cd. Juarez Chih., Mexico.
RECALLED BY
Tipp Enterprises, doing business as Novamex, El Paso, Texas, by telephone and fax on or about April 30, 1999. Firm-initiated recall complete.
DISTRIBUTION
Nationwide.
QUANTITY
100,000 cases (24 per case) were distributed.
REASON
Product failed to bear an ingredient statement.
________
PRODUCT
Tuv Taam brand Cole Slaw, in 14-ounce rigid plastic containers. Recall #F-436-0.
CODE
All codes.
MANUFACTURER
Tuv Taam Corporation, Brooklyn, New York.
RECALLED BY
Manufacturer, by visit starting on December 8, 1999. Completed recall resulted from sample analysis and follow by the New York State Department of Agriculture and Markets.
DISTRIBUTION
New York and New Jersey.
QUANTITY
Approximately 15,000 packages were distributed.
REASON
Product contained undeclared potassium sorbate.
________
PRODUCT
Cypress Medical Products Sterile Procedure Kits and Tray containing antiseptic products recalled by Clinipad Corporation. Recall #Z-522-0.
CODE
a) Kits containing Clinidine Ointment, Sterile, 1.0 gram
I.V. Start Kits:
Reorder #82-03, Economy, 50 kits/case, lot 80410, and all lots starting with "9" or "0"
Reorder #82-05, Deluxe, 50 kits/case, all lots starting with "9" or "0"
Reorder #82-06, Latex-free, 50 kits/case, all lots starting with "9" or "0"
Reorder #82-07, Premium, 50 kits/case, lots 71103, 71104, 80409, all lots starting with "9" or "0".
Daily Catheter Care Trays, 20 trays/case, reorder #86-101, lot 71205, 71208, 90403, 90225.
b) Kits containing Clinidine Ointment - 4339SC, Clinguard
Protective Dressing Prep Pad - 8133S-B, Cliniswab
Isopropyl Alcohol Antiseptic Swabsticks - 1245SB, and Clinipad Alcohol Prep - 0110:
Omnicare Central Line Dressing Kit, reorder #84-117, lots 71215 and 80404;
c) Kits containing Clinipad Alcohol Prep - 0110:
Sharp Debridement Package, 20 kits/case, order number SDP, reorder #84-160, lots 70812 and 70813;
d) Kits containing Clinipad Iodophor PVP Iodine USP Prep Pads - 9153:
Skin Staple Remover kits:
Reorder #82-15, Premium, 50 kits/case, lot 71016
Reorder #1295-815, Starline, 50 kits/case, lot 70903
Reorder #241, PSS Select, 50 kits/case, lot 80214
Suture Removal Kit, reorder #81-213, lot 80305.
MANUFACTURER
Cypress Medical Products, Ltd., McHenry, Illinois.
RECALLED BY
Manufacturer, by letter on March 22, 2000. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide.
QUANTITY
96,736 kits/trays were distributed.
REASON
The antiseptic products may be contaminated with bacteria.
________
PRODUCT
Various Sterile Procedure Kits containing antiseptic products recalled by Clinipad Corporation:
ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with
T-Fasteners, Complete Kit, 14 French (List #00227);
ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with
T-Fasteners, Complete Kit, 18 French (List #00238);
ROSS Flexiflo Introducer Gastrostomy Kit with Brown/Mueller
T-Fastener Set, Complete Kit, 18 French (List #50190);
Ross Inverta-PEG Pull Technique Gastrostomy Kit with Roll-Tip Bumper, Complete Kit with Snare, 20 French (List #52002);
Ross Inverta-PEG Over-the-Guidewire Gastrostomy Kit with Roll-Tip Bumper, Complete Kit with Snare, 20 French (List #52354);
ROSS Flexiflow Versa-PEG Gastrostomy Kit, Complete Kit, 22 French (list #50520);
ROSS Flexiflo Inverta-PEG Over-the-Guidewire Gastrostomy Kit with Roll-Tip Bumper, Complete Kit with Snare, 20 French (List #51492) [Obsolete]. Recall #Z-523-0.
CODE
All lot numbers beginning with 25 through 62.
MANUFACTURER
Ross Products Division, Abbott Laboratories, Inc., Columbus, Ohio.
RECALLED BY
Manufacturer, by letters dated March 20 and 27, 2000, and providing stickers to be placed on kits advising users not to use the Clinipad products and to destroy them. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide, Chile, Hungary, Korea, Mexico, Netherlands, Singapore, Turkey.
QUANTITY
77,492 (including the Class II portion).
REASON
Kits contain Clinipad antiseptic products which may be contaminated with bacteria.
________
PRODUCT
Sterile Procedure Kits containing antiseptic products recalled by Clinipad Corporation:
ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with Reinforced Webbed Bumper, Complete Kit, 16 French (List #50724); ROSS Flexiflo Over-the-Guidewire Gastrostomy Kit with Reinforced Webbed Bumper, Complete Kit, 20 French (List #50738). Recall #Z-523-0.
CODE
All lot numbers beginning with 25 through 62.
MANUFACTURER
Ross Products Division, Abbott Laboratories, Inc., Columbus, Ohio.
RECALLED BY
Manufacturer, by letters dated March 20 and 27, 2000, and providing stickers to be placed on kits advising the users not to use the Clinipad products and to destroy them. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide, Chile, Hungary, Korea, Mexico, Netherlands, Singapore, Turkey.
QUANTITY
77,492 (including the Class I portion).
REASON
Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
Heart Valve Allograft Models AVOO, PVOO, AVO5.
Recall #Z-628/631-0.
CODE
Serial Numbers: 3887502, 3887391, 6100597, 6276377.
MANUFACTURER
Cryolife, Inc., Kennesaw, Georgia.
RECALLED BY
Manufacturer, by letter on March 23 and 27, 2000. Firm-initiated recall complete.
DISTRIBUTION
Tennessee, Oklahoma, Michigan.
QUANTITY
4 valves.
REASON
Donor did not meet current guidelines regarding serodilution of plasma.
________
PRODUCT
Hemodialysis/Apheresis Trays containing recalled Clinipad products.
Recall #Z-632-0.
CODE
All items listed below are Vas-Cath products.
5500102 |
Soft-Cell Catheterless Tray |
5500103 |
Soft-Cell Tray 12.5F x 19cm |
5500114 |
Flexxicon II Catheterless Tray |
5500125 |
Soft-Cell Tray 12.5F x 27cm |
550012 |
Soft-Cell Tray 12.5F x 15cm |
5500130 |
Soft-Cell Tray 12.5F x 17cm |
5500132 |
Soft-Cell Tray 12.5F x 27cm |
5525150 |
Flexxicon II Tray 12F x 15cm |
5525200 |
Flexxicon II Tray 12F x 20cm |
5525240 |
Flexxicon II Tray 12F x 24cm |
5526120 |
Flexxicon II 12.5 cm Tray |
5526150 |
Flexxicon II Tray 12F x 15cm |
5526200 |
Flexxicon II Tray 12F x 20cm |
5526620 |
Flexxicon II Tray 12F x 12.5cm |
5526650 |
Flexxicon II Tray 12F x 15cm |
5526700 |
Flexxicon II Tray 12F x 20cm |
5535120 |
Soft-Cell Tray 12.5 x 12cm |
5535190 |
Soft-Cell Tray 12.5F x 19cm |
5535230 |
Soft-Cell Tray 12.5F x 23cm |
5535620 |
Soft-Cell Tray 12.5F x 12cm |
5535690 |
Soft-Cell Tray 12.5F x 19cm |
5535730 |
Soft-Cell Tray 12.5F x 23cm |
5536120 |
Soft-Cell Tray 12.5F x 12cm |
5536190 |
Soft-Cell Tray 12.5F x 19cm |
5536230 |
Soft Cell Pre-Curved Dual Lumen Chronic Catheter |
5545150 |
Vaccess Tray 8F x 15cm |
5545200 |
Vaccess Tray 8F x 20cm |
5545700 |
Vaccess Tray 8F x 20cm |
5591000 |
Niagara Catheterless Tray |
5595150 |
Niagara Tray 13.5F x 15cm |
5595200 |
Niagara Tray 13.5F x 20cm |
5595240 |
Niagara Tray 13.5F x 24cm |
5596120 |
Niagara Tray 13.5F x 12.5cm |
5596150 |
Niagara Tray 13.5F x 15cm |
5596200 |
Niagara Tray 13.5F x 20cm |
6531000 |
Opti-Flow Catheterless Tray |
5635192 |
Opti-Flow Tray 14.5F x 19cm |
5635692 |
Opti-Flow Tray 14.5F x 19cm |
5635732 |
Opti-Flow Tray 14.5F x 23cm |
5635232 |
Opti-Flow Tray 14.5F x 23cm |
5636192 |
Opti-Flow Tray 14.5F x 19cm |
5636232 |
Opti-Flow Tray 14.5F x 23cm. |
MANUFACTURER
Bard Access Systems, Inc., Salt Lake City, Utah.
RECALLED BY
Manufacturer, by letters on March 20 and 22, 2000. Customers were given the option of returning any affected product for credit or removing the Clinipad components. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad, Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide.
QUANTITY
108,347 kits/trays were distributed.
REASON
Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
LSL Sterile Mid-Stream Urine Specimen Kit, 120 ml, containing recalled Clinipad product. Recall #Z-633-0.
CODE
Kit Reorder #LSL1030, All lots beginning with 7, 8, 9, and 0A3006, 0B3007.
MANUFACTURER
LSL Industries, Inc., Chicago, Illinois.
RECALLED BY
Manufacturer, by letters dated March 20, 1000, and April 6, 2000, instructing users to hold kits in quarantine until stickers were provided to be placed on the kits advising the user not to use the Clinipad product and to destroy it. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Tennessee, Kansas, Vermont, Texas, Michigan, Pennsylvania, New York, Oklahoma, Massachusetts, Indiana, Florida, New Jersey, Oregon, Arizona, Alaska.
QUANTITY
64,400 kits were distributed.
REASON
Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
Deltec Procedural Tray (containing recalled Clinipad products) for use with CliniCath Peripherally Inserted Catheters, Product Numbers 21-2501-01 and 21-2515-01; Deltec CliniCath Peripherally Inserted Catheters, with Tray, Product Numbers 21-2508-01, 21-2509-01, 21-2510-01, 21-2511-01, 21-2512-01, and 21-2513-01. Recall #Z-634-0.
CODE
All lots distributed since 1/1/97.
MANUFACTURER
Sims Deltec, Inc., St. Paul, Minnesota.
RECALLED BY
Manufacturer, by letter dated April 5, 2000. Firm stated that consignees could keep CliniCath products needed for urgent patient needs and return the rest. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide and international.
QUANTITY
3,512 trays were distributed.
REASON
Trays contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
Zipprep Electrode Starter Kits containing Clinipad Alcohol Antiseptic Swabsticks, intended for use with EEG monitoring Systems. Recall #Z-635-0.
CODE
Aspect Product #185-0043; Aspect Lot Numbers: 418442, 413659, 423659, 417506.
MANUFACTURER
Aspect Medical Systems, Inc., Natick, Massachusetts.
RECALLED BY
Manufacturer, by notifying customers on March 29, 2000 to discontinues and destroy any products on hand. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Washington state, Florida, West Virginia, California, Ohio, Massachusetts, Pennsylvania, Louisiana, Kentucky, and international.
QUANTITY
81 kits (all past expiration dating).
REASON
Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
General Procedure Kit. Recall #Z-637-0.
CODE
Catalog DYNJTS0300, Lot 9904007
DYNJTS4302, Lot 9904002
DYNJTS0400, Lot 9903041
DYNJTS4306, Lot 9903037
DYNJTS7025, Lot 9903035
DYNJTS4301, Lot 9904001
DYNJTS4303, Lot 9903036.
MANUFACTURER
Physician Industries Manufacturing, West Valley City, Utah.
RECALLED BY
Manufacturer, by telephone on May 6, 1999. Firm-initiated recall complete.
DISTRIBUTION
Georgia, Illinois, Nevada, Pennsylvania, Texas.
QUANTITY
2,880 kits were distributed.
REASON
Distributed kits had positive bioburden test results after EtO sterilization.
________
PRODUCT
Allograft Heart Valve, Model AVOO and PVOO.
Recall #Z-638/639-0.
CODE
Serial Numbers: 3887693, 3887707, 3895343, 3895351, 6004946, 6004954, 6262155, 6268331, 6297846.
MANUFACTURER
Cryolife, Inc., Kennesaw, Georgia.
RECALLED BY
Manufacturer, by letters on April 3 and 6, 2000. Firm-initiated recall complete.
DISTRIBUTION
New York, Colorado, Tennessee, Florida, Michigan, Maryland, Ohio, California.
QUANTITY
9 valves.
REASON
Donor did not meet current guideline regarding serodilution of plasma.
________
PRODUCT
Hermitage Eye Pads, Sterile 2 1/8" X 2 5/8", Product of China. Sterilized in the U.S.A, for use as a bandage over the eye for protection or absorption of secretions: a) Item No. EP-1600; b) Item No. EP-2000. Recall #Z-640/641-0.
CODE
Lot #5588.
MANUFACTURER
Shaoxing Life Surgical Dressing Company, Ltd., Zhejiang Prov. Shaoxing County, China.
RECALLED BY
Hermitage Hospital Products, Inc., Niantic, Connecticut, by telephone on April 19, 2000, followed by fax. Firm-initiated recall ongoing.
DISTRIBUTION
Arkansas, California, Oklahoma, Puerto Rico.
QUANTITY
a) 33 cases (600 per case); b) 33 cases (2000 per case).
REASON
Product was not sterilized and incorrectly transferred to released inventory without being gamma sterilized.
________
PRODUCT
Allograft Heart Valve, Models PV00, AV00, AV40, MV10 PV05.
Recall #Z-642/646-0.
CODE
Serial Numbers: 3889149, 3911446, 3911454, 6965792, 6965768, 6313744.
MANUFACTURER
Cryolife, Inc., Kennesaw, Georgia.
RECALLED BY
Manufacturer, by letters on April 7, 12, and 13, 2000. Firm-initiated recall complete.
DISTRIBUTION
New York, Pennsylvania, Louisiana, California, North Carolina, Canada.
QUANTITY
6 valves.
REASON
Donors did not meet current guidelines regarding serodilution of plasma.
________
PRODUCT
Disposable Coronary Control Syringe, 20mL,
Catalog #K01-04072. Recall #Z-647-0.
CODE
Lot #A140030.
MANUFACTURER
Merit Medical Systems, Inc., South Jordan, Utah.
RECALLED BY
Manufacturer, by telephone on January 21, 2000. Firm-initiated recall complete.
DISTRIBUTION
South Carolina.
QUANTITY
25 syringes.
REASON
Syringes were shipped to customer prior to terminal sterilization processing.
________
PRODUCT
Flo-Rester Sizer Pack (of blood vessel sizers), used to determine the internal diameters of arteries and veins:
Catalog Numbers: a) FRSZPK05; b) FRSZ400; c) FRSZ300;
d) FRSZ250; e) FRSZ200; f) FRSZ150. Recall #Z-648/653-0.
CODE
Lot Numbers: a) 526069, 525165, 523878, 522704, 521864, 518549, and 515997; b) 526067, 525588, 524943, 520805, and 516003;
c) lot number 516002; d) 516001; e) 516000; f) 515999.
MANUFACTURER
Bio-Vascular, Inc., St. Paul, Minnesota.
RECALLED BY
Manufacturer, by letter dated April 19, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide, Brazil, Canada, Denmark, Germany, United Kingdom.
QUANTITY
107 sizers were distributed.
REASON
The bulb of the device may loosen and become detached from the shaft.
________
PRODUCT
VACUTAINER Urine Collection Kit for Midstream Specimens containing recalled Clinipad products.
Recall #Z-659-0.
CODE
Catalog Number 364952 Vacutainer Urine Collection kit with
sterile cup, 5ml preservative tube, and Clinipad Rantex
Wipe, Lot 9E165, 9F229, 9H142, 9J205, 9J206;
Catalog Number 364972 Vacutainer Urine Collection kit with
sterile cup, 12 ml conical urinalysis tube, 5 ml
preservative tube, and Clinical Rantex Wipe, Lot 9E170,
9E171, 9F233, 9G085, 9H147, 9J201.
MANUFACTURER
Becton Dickinson & Company, Broken Bow, Nebraska.
RECALLED BY
Becton Dickinson & Company, Franklin Lakes, New Jersey, by letter dated April 7, 2000, instructing customers to place warning stickers on kits. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide.
QUANTITY
1,428,050 kits were distributed.
REASON
Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
FLOW-20, used as a gastrointestinal feeding tube:
a) Flow-20-Push; b) Flow-20-Pull; c) Flow-20-Pull-S.
Recall #Z-660/662-0.
CODE
Lot Numbers: a) 1276176 and 1280434; b) 1276560;
c) 1277439 and 1278396.
MANUFACTURER
Wilson-Cook Medical, Inc., Winston-Salem, North Carolina.
RECALLED BY
Manufacturer, by telephone, fax and mail beginning November 10, 1999. Firm-initiated recall complete.
DISTRIBUTION
Alabama, Arizona, California, Connecticut, Florida, Louisiana, Massachusetts, Missouri, New Hampshire, New York, North Dakota, Ohio, Pennsylvania, Texas, Vermont, Wisconsin, Australia, Hungary, Jordan, Singapore.
QUANTITY
246 units were distributed.
REASON
Kits may contain a larger external bumper, which may not properly secure the feeding tube.
________
PRODUCT
Magnetom 42 SP Magnetic Resonance Imaging System with
Software Version Number 2A2.7B. Computer aided medical device generates magnetic field resulting in a digital image of the body:
a) Magnetom 42 SP Magnetic Resonance Imaging Systems, Model No. 89 69 131 (SP-A);
b) Magnetom 42 SP Magnetic Resonance Imaging Systems, Model No. 87-29-485 (SP). Recall #Z-664/665-0.
CODE
Each unit is identified with a unique serial number. Those units sold (194 units) in the USA with the software version 2A2.7B are listed as follows:
03010 |
03011 |
03014 |
03015 |
03020 |
03022 |
03023 |
||||||
03026 |
03027 |
03034 |
03036 |
03047 |
03048 |
03054 |
||||||
03055 |
03058 |
03059 |
03060 |
03062 |
03063 |
03064 |
||||||
03066 |
03067 |
03068 |
03069 |
03070 |
03073 |
03084 |
||||||
03088 |
03091 |
03094 |
03095 |
03096 |
03097 |
03099 |
||||||
03102 |
03103 |
03115 |
03116 |
03120 |
03121 |
03125 |
||||||
03130 |
03139 |
03141 |
03144 |
03146 |
03148 |
03155 |
||||||
03157 |
03158 |
03160 |
03164 |
03168 |
03169 |
03179 |
||||||
03185 |
03186 |
03188 |
03189 |
03202 |
03203 |
03205 |
||||||
03209 |
03212 |
03213 |
03215 |
03218 |
03222 |
03227 |
||||||
03228 |
03229 |
03230 |
03231 |
03234 |
03235 |
03236 |
||||||
03237 |
03238 |
03239 |
03240 |
03242 |
03245 |
03256 |
||||||
03257 |
03258 |
03261 |
03262 |
04004 |
04015 |
04016 |
||||||
04017 |
04019 |
04020 |
04026 |
04027 |
04028 |
04031 |
||||||
04033 |
04036 |
04037 |
04038 |
04041 |
04042 |
04044 |
||||||
04045 |
04046 |
04047 |
04049 |
04052 |
04053 |
04058 |
||||||
04059 |
04063 |
04064 |
04068 |
04070 |
04076 |
04080 |
||||||
04082 |
04083 |
04084 |
04087 |
04094 |
04095 |
04096 |
||||||
04099 |
04102 |
04105 |
04107 |
04108 |
04114 |
04115 |
||||||
04117 |
04119 |
04121 |
04126 |
04127 |
04128 |
04136 |
||||||
04137 |
04142 |
04147 |
04153 |
04154 |
04156 |
04157 |
||||||
04160 |
04162 |
04163 |
04164 |
04166 |
04167 |
04169 |
||||||
04170 |
04171 |
04172 |
04173 |
04176 |
04180 |
04189 |
||||||
04190 |
04192 |
04201 |
04204 |
04208 |
04209 |
04210 |
||||||
04211 |
04213 |
04214 |
04215 |
04218 |
04220 |
04229 |
||||||
04238 |
04244 |
04248 |
04258 |
04259 |
04265 |
04266 |
||||||
04268 |
04273 |
04274 |
04277 |
04278 |
04279 |
04280 |
||||||
04285 |
04289 |
04290 |
04292 |
04296 |
||||||||
Those units sold overseas (193 units) are with the affected software version are listed below: |
||||||||||||
03006 |
03013 |
03018 |
03019 |
03024 |
03025 |
03031 |
||||||
03033 |
03039 |
03042 |
03045 |
03046 |
03049 |
03051 |
||||||
03057 |
03061 |
03071 |
03072 |
03078 |
03080 |
03081 |
||||||
03082 |
03083 |
03086 |
03093 |
03108 |
03109 |
03111 |
||||||
03112 |
03113 |
03114 |
03118 |
03119 |
03122 |
03124 |
||||||
03126 |
03127 |
03129 |
03131 |
03132 |
03133 |
03135 |
||||||
03143 |
03145 |
03147 |
03151 |
03153 |
03154 |
03165 |
||||||
03166 |
03170 |
03171 |
03173 |
03182 |
03183 |
03184 |
||||||
03187 |
03190 |
03191 |
03208 |
03217 |
03224 |
03226 |
||||||
03244 |
03246 |
03260 |
04002 |
04003 |
04005 |
04007 |
||||||
04008 |
04012 |
04013 |
04018 |
04021 |
04023 |
04024 |
||||||
04025 |
04029 |
04030 |
04032 |
04039 |
04040 |
04043 |
||||||
04048 |
04050 |
04055 |
04056 |
04060 |
04061 |
04062 |
||||||
04065 |
04066 |
04069 |
04071 |
04072 |
04073 |
04074 |
||||||
04075 |
04077 |
04078 |
04079 |
04086 |
04088 |
04090 |
||||||
04092 |
04093 |
04097 |
04100 |
04104 |
04109 |
04110 |
||||||
04111 |
04113 |
04118 |
04124 |
04125 |
04129 |
04130 |
||||||
04131 |
04132 |
04133 |
04134 |
04139 |
04143 |
04144 |
||||||
04146 |
04148 |
04149 |
04150 |
04152 |
04158 |
04175 |
||||||
04177 |
04178 |
04179 |
04181 |
04183 |
04184 |
04185 |
||||||
04186 |
04191 |
04193 |
04194 |
04195 |
04197 |
04198 |
||||||
04199 |
04200 |
04205 |
04206 |
04216 |
04217 |
04221 |
||||||
04223 |
04224 |
04225 |
04228 |
04233 |
04234 |
04235 |
||||||
04236 |
04246 |
04247 |
04249 |
04250 |
04251 |
04252 |
||||||
04253 |
04254 |
04255 |
04256 |
04257 |
04260 |
04261 |
||||||
04262 |
04263 |
04264 |
04269 |
04271 |
04272 |
04275 |
||||||
04276 |
04282 |
04283 |
04288 |
04291 |
04294 |
04295 |
||||||
04301 |
04302 |
04303 |
04304. |
MANUFACTURER
Siemens AG, Erlangen, Germany.
RECALLED BY
Siemens Medical Systems, Inc., Iselin, New Jersey, by letter dated February 24, 2000. Firm-initiated field correction ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
Approximately 387 units.
REASON
Patients who are registered on the MRI systems with birth dates of January 1, 2000 or later will be misinterpreted by the software to have been born in 1900.
________
PRODUCT
Product insert for Acetaminophen (Enzymatic Rate) Reagent, used for the quantitative determination of Acetaminophen in serum, Catalog #502-13. Recall #Z-667-0.
CODE
Catalog No. 502-13.
MANUFACTURER
Diagnostic Chemicals Ltd., Charlottestown, PE, Canada.
RECALLED BY
Manufacturer, by letters dated March 15 and 21, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and Canada
QUANTITY
310 units were distributed.
REASON
The "Product Insert" does not contain information on interference from N-acetylcysteine (NAC), the drug of choice for treating acetaminophen overdose.
_______
UPDATE
Abbott ARCHITECT i System with Software Version 1.0 Processing Module, List No. 08C89-00-1, designed to perform automated immunoassay tests bassed on the use of Chemiluminescent Microparticle Immunoassay (CMIA) detection technology. Recall #Z-611-0, which appeared in the May 3, 2000 Enforcement Report should read: RECALLED BY: Manufacturer, by telephone on February 25, 1999.
________
PRODUCT
GAP IgG Helicobacter Pylori Enzyme Immunoassay Kit, for the qualitative detection of IgG specific antibodies to helicobacter pylori in human serum. Recall #Z-610-0.
CODE
Catalog #404 2050, Lot #0426753-1/2 EXP 11/00.
MANUFACTURER
Biomerica, Inc., Newport Beach, California.
RECALLED BY
Manufacturer, by telephone on or about March 17, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
California.
QUANTITY
252 kits were distributed.
REASON
The stability parameters were starting to draft and would be out-of-specification before the expiration date.
________
PRODUCT
Endoscopic Blunt Dissecting Instrument, used to probe or move organs during surgery. Recall #Z-636-0.
CODE
KJJU02, KJJU03, BKJU01, BKJU02.
MANUFACTURER
Manufacturas Frontera Norte SA de CV, CD Juarez, Mexico.
RECALLED BY
Eagle Endoscopic, Inc., Bartonville, Texas, by telephone on April 14, 2000, and by telephone followed by fax on April 17, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
Illinois and Tennessee.
QUANTITY
473 cases (20 units per case) were distributed.
REASON
Tip separated from body while in use.
________
PRODUCT
Merge ExamWorks Video Units, used with equipment for connecting diverse diagnostic imaging products: Model/Catalog Numbers: 006-04100-00, 006-04100-01, 006-04100-00S, 006-04100-01S, and 006-04100-02. Recall #Z-654/658-0.
CODE
All units shipped prior to March 30, 2000.
MANUFACTURER
Nicolet, Madison, Wisconsin.
RECALLED BY
Merge Technologies, Inc., West Allis, Wisconsin, by "Important Customer Product Advisory" dated April 21, 2000. Firm-initiated field correction ongoing.
DISTRIBUTION
California, Connecticut, Kansas, Minnesota, Missouri, Nebraska, Ohio, Oregon, Virginia, Wisconsin, Brazil, Netherlands, Sweden.
QUANTITY
103 units were distributed.
REASON
There is a potential for horizontal banding to appear on video screen images and images printed on laser film.
________
PRODUCT
Kodak Oncology Film, identified as ECFLM film, 35 x 43 cm, Catalog No. 8715757. Recall #Z-663-0.
CODE
Emulsion Numbers: 0116-022-12, 0116-022-14, 0116-022-16.
MANUFACTURER
Eastman Kodak Company, Windsor, Colorado.
RECALLED BY
Eastman Kodak Company, Rochester, New York, by telephone on March 24, 2000, and letters dated April 13, 2000. Firm-initiated recall ongoing.
DISTRIBUTION
California, Missouri, North Carolina, Pennsylvania, Texas, Wisconsin.
QUANTITY
26 boxes (100 sheets per box) were distributed.
REASON
Graphic Arts film was slit, chopped, and packaged as Oncology X-ray Film.
________
PRODUCT
Ampicillin/Sulbactam 10/10 mcg Antimicrobial Susceptibility Test Disk, packaged 50 disks per box, intended for use in the standardized disk susceptibility test and is the recommended method of performing antimicrobial disk diffusion testing. Recall #Z-666-0.
CODE
Product #33-135, Lot #216785 EXP 4/02.
MANUFACTURER
Oxoid Ltd., Basingstoke Hants RG248PW, United Kingdom.
RECALLED BY
Remel Inc., Lenexa, Kansas, by telephone during March 13 - 15, 2000. Firm-initiated recall complete.
DISTRIBUTION
Nationwide.
QUANTITY
100 packages were distributed.
REASON
The zones of inhibition obtained were below the acceptable zone diameter of the NCCLS Standard for Antimicrobial Disk Susceptibility Tests.
________
PRODUCT
Bovine mastitis treatment kits containing Clinipad alcohol pads:
a) U.S. Vet Masti-Clear Penicillin G Procaine in Sesame Oil Lactating Cow Mastitis treatment. GC Hanford - firm on label.
b) U.S. Vet go-dry. Penicillin G Procaine in Sesame Oil Dry Cow Mastitis Treatment - GC Hanford - firm on label
c) First Choice Dry Cow Mastitis Treatment, Penicillin G Procaine in Sesame Oil - KenAG Inc., Ashland , O H- firm on label
d) Aqua-Mast II Lactating Cow Mastitis Treatment Penicillin G Procaine in Sesame Oil. KenAG, Inc., Ashland, OH - firm on label
e) Amoxi-Mist Lactating Cow Formula - Rx veterinary product - Pfizer Animal Health, Exton, PA - firm on label
f) Dariclox Lactating Cow Formula - Rx veterinary product - Pfizer Animal Health Exton, PA - firm on label
g) Orbenin Quick Release Cloxacillin Sodium Intramammary Infusion USP.Veterinary product - Canadian market - SmithKline Beecham Animal Health, Mississauga, Ontario - firm on label.
h) Orbenin-DC Dry Cow Intramammary Infusion - Rx veterinary product - Pfizer Animal Health, Exton, PA firm on label.
Recall #V-082/089-0.
CODE
All lots produced from 3/7/97 to 3/9/2000 - Kits produced AFTER 3/9/2000 will have Webcol Alcohol Preps.
MANUFACTURER
G.C. Hanford Manufacturing Company, Syracuse, New York.
RECALLED BY
Manufacturer, by letter mailed on March 29, 2000, requesting consignees to remove and destroy all Clinipad alcohol preps packaged with the kits. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide.
QUANTITY
619,820 cartons were distributed.
REASON
Kits contain Clinipad products labeled as sterile for which Clinipad is unable to assure the sterility.
________
PRODUCT
a) Albadry Plus Suspension Kits - Penicillin G procaine and novobiocin sodium dry cow formula. 12-10 ml syringes and clinipad alcohol pads for intramammary infusion for treatment of mastitis;
b) Albadry Plus Suspension Kits - Penicillin G procaine and novobiocin sodium dry cow formula. 144-10 ml syringes and clinipad alcohol pads for intramammary infusion for treatment of mastitis. Recall #V-090/091-0.
CODE
All lots - Clinipad lot numbers packaged in kits - 705751, 801134, 803577, 803755.
MANUFACTURER
Pharmacia & Upjohn, Kalamazoo, Michigan.
RECALLED BY
Manufacturer, by letter mailed on April 6, 2000. Firm-initiated recall ongoing.
COMPONENT MANUFACTURER
Clinipad, Corporation, Rocky Hill, Connecticut.
DISTRIBUTION
Nationwide.
QUANTITY
176,000 cartons of 12 and 5,850 pails of 144 were distributed.
REASON
Kits contain Clinipad Alcohol Pads labeled as sterile. Clinipad is unable to assure sterility.
________
PRODUCT
Abbott Ambulatory Infusion Manager Plus Single-Channel Infusion Pump, five types of programming options available for TPN, pain management, intermittent, variable time and contiuous delivery of medications and nutritional fluids.
Safety Alert #N-014-0.
CODE
List #13967, Serial Numbers 12300001/12320000.
MANUFACTURER
Abbott Laboratories, Inc., San Diego, California.
ALERTED BY
Abbott Laboratories, Inc., Abbott Park, Illinois, by letter date, by letter on March 2, 2000.
DISTRIBUTION
Nationwide.
QUANTITY
15,947 units were distributed.
REASON
There is a possibility that medication therapy may be missed on March 1, 2000, due to the leap year, if the pumps were not appropriately reprogrammed on March 1, 2000.
END OF ENFORCEMENT REPORT FOR MAY 10, 2000
####
FDA News Page |   FDA Home Page
Office of Public Affairs