CODE
All with sell-by date prior to 11/12/97.
MANUFACTURER
Kohl's Food Stores, Division of A&P Tea Company, Inc., doing
business as Kohl's Bakery & Deli Kitchens, Milwaukee,
Wisconsin.
RECALLED BY
Manufacturer, by telephone on November 8, 1997. Firm-initiated
recall complete.
DISTRIBUTION
Wisconsin.
QUANTITY
600 cakes were for sale. All were relabeled with corrected
labels.
REASON
Product contained undeclared eggs.
PRODUCT
Safeway Select Pralines'N Cream Light Ice Cream, in half gallon
containers. Recall #F-537-8.
CODE
All codes.
MANUFACTURER
Bellevue Ice Cream, Bellevue, Washington; Capitol Heights Ice
Cream, Capitol Heights, Maryland, and Von's, Los Angeles,
California.
RECALLED BY
Safeway, Inc., Pleasanton, California, by E-mail and telephone on
May 19, 1998, and by press release on May 19, 1998.
Firm-initiated recall complete.
DISTRIBUTION
Alaska, Washington state, Oregon, Idaho, Montana, California,
Nevada, Hawaii, Maryland, and District of Columbia.
QUANTITY
1,512 cases containing six 1/2 gallon containers were
distributed.
REASON
Product contains undeclared pecans.
PRODUCT
Duncan Hines Chewy Fudge Brownie Mix, family size, packed in 21
ounce (1 pound 5 ounce) 595 gram cardboard retail packages.
Recall #F-538-8.
CODE
8112M5
MANUFACTURER
Olmarc Packaging Corporation, North Lake, Illinois.
RECALLED BY
Aurora Foods, Inc.,Columbus, Ohio, by letter, press release and
notified the Food Allergy Network on May 28, 1998.
Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide.
QUANTITY
8,942 cases (107,304 retail packages) were distributed.
REASON
Product contains undeclared walnuts.
PRODUCT
La Martinique Blue Cheese Vinaigrette Restaurant Dressing, in 8
fluid ounce (236 ml) containers. Recall #F-540-8.
CODE
"AUG 16 99".
MANUFACTURER
Reily Foods/Blue Plate Foods, New Orleans, Louisiana.
RECALLED BY
Reily Foods Company, Inc., New Orleans, Louisiana, by letter sent
on April 14, 1998, and by press release on April 30, 1998.
Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide.
QUANTITY
2,805 cases (12 units per case) were distributed.
REASON
Product is contaminated with Listeria monocytogenes.
CODE
HA04098.
MANUFACTURER
ConAgra, Inc., Hastings, Minnesota.
RECALLED BY
ConAgra Flour Milling Company, Omaha, Nebraska, by telephone on
April 23, 1998, and June 10, 1998. Firm-initiated recall
complete.
DISTRIBUTION
Colorado, Maryland, New Jersey, New York, Ohio.
QUANTITY
182 50-pound bags were distributed.
REASON
Product contains undeclared wheat flour.
PRODUCT
Oxygen, Medical, USP, Compressed, in high pressure E cylinders,
Rx. Recall #D-182-8.
CODE
Lot numbers: 42001A and 42001B.
MANUFACTURER
Sieps Homecare, Inc., doing business as Family Homecare, Rock
Rapids, Iowa.
RECALLED BY
Manufacturer, by telephone on May 27, 1998. Firm-initiated
recall complete.
DISTRIBUTION
Minnesota.
QUANTITY
2 cylinders were distributed.
REASON
Subpotent and good manufacturing practice deviations.
CODE
E-size aluminum cylinders with lot # sticker as follows:
04259803-3 units; 01139801-3 units; 01139802-3 units;
01139803-3 units; 01139804-3 units; 01139805-3 units;
01229803-3 units; 01229806-3 units; 01229807-3 units;
01229808-3 units; 01159801-3 units; 01159802-3 units;
01159803-3 units; 01159804-3 units; 01159805-3 units;
01159806-3 units; 01159808-3 units; 01159809-3 units;
01059801-3 units; 01059802-3 units; 01059803-3 units;
01059804-3 units; 01059805-3 units; 01059806-3 units;
01059807-3 units; 12299701-3 units; 12299702-3 units;
12299703-3 units; 12299704-3 units; 10149702-3 units;
10109702-3 units; 10109703-3 units; 10109705-3 units;
10109706-3 units; 10109707-3 units; 10109708-3 units;
10089701-3 units; 10089702-3 units; 10089704-3 units;
10089708-3 units; 10069701-3 units; 10069702-2 units;
10069703-3 units; 10069705-3 units; 10039701-3 units;
10039704-3 units; 10039705-3 units; 10019704-3 units;
10019705-3 units; 10019706-3 units; 09309701-3 units;
09309702-3 units; 09309707-3 units; 09309708-3 units;
09309709-3 units; 09309711-3 units; 09199701-3 units;
09199702-3 units; 09199704-3 units; 09179701-3 units;
09179707-3 units; 09179708-3 units; 09179709-3 units;
08139701-3 units; 08129701-3 units; 08129702-3 units;
08129703-3 units; 08129704-3 units; 08129705-3 units;
08129706-3 units; 08059702-3 units; 08059703-3 units;
08059705-3 units
M6-size aluminum cylinders with lot # sticker as follows: 02279808-3 units; 01229801-3 units; 01229802-3 units; 01229804-3 units; 01229805-3 units; 01159807-3 units; 01029801-3 units; 01029802-3 units; 12299705-3 units; 12299706-3 units; 10109701-3 units; 10109704-2 units; 10109709-3 units; 10089703-1 unit; 10089705-3 units; 10089706-3 units; 10089707-1 unit; 10089709-3 units; 10069704-3 units; 10069706-3 units; 10039702-3 units; 10039703-3 units; 10039706-3 units; 10039707-3 units; 10039708-1 unit; 10019701-3 units; 10019702-3 units; 10019703-3 units; 09309703-3 units; 09309704-3 units; 09309705-3 units; 09309706-3 units; 09309710-3 units; 09309712-3 units; 09199703-3 units; 09179702-3 units; 09179703-3 units; 09179704-3 units; 09179705-3 units; 09179706-3 units; 08139702-3 units; 08139703-3 units; 08059701-3 units; 08059704-3 units
MANUFACTURER
Shiv Respiratory Equipment, Inc., doing business as American
Respiratory Equipment, Inc., Latham, New York.
RECALLED BY
Manufacturer, by memorandum dated May 17, 1998. Firm-initiated
recall ongoing.
DISTRIBUTION
New York.
QUANTITY
218 E-size cylinders; 125 M6-size cylinders were distributed.
REASON
Subpotency.
UPDATE
SureStep Consumer Blood Glucose Meters (LifeScan, Milpitas,
California), Recall #Z-631-8, which appeared in the July 1, 1998
Enforcement Report should read: CODE: All Models, Serials #'s:
L6000 XX XXXXX through L7205 XX XXXXX and L7206 GA 00001
through L7206 GA 01128
Ohmeda Tec 6 Desflurane Vaporizers (Note: Excludes North American Drager (NAD) and Dragerwerk Tec 6 Variant Vaporizers). Potentially affected vaporizers would include: Model Part Number Tec 6 120v-USA 1107-9001-000 Tec 6 120v-Canada 1107-9012-000 Tec 6 120v-France 1107-9003-000 Tec 6 220/240v-England 1107-9002-000 Tec 6 220/240v-Germany 1107-9004-000 Tec 6 220/240v-Italy 1107-9005-000 Tec 6 220/240v-Dutch 1107-9006-000 Tec 6 220/240v-Sweden 1107-9007-000 Tec 6 220/240v-Spain 1107-9008-000 Tec 6 220/240v-Finland 1107-9009-000 Tec 6 220/240v-Norway 1107-9010-000 Tec 6 220/240v-Danish 1107-9011-000 Tec 6 220/240v-England-Europe 1107-9014-000 Vaporizers are desflurane anaesthetic vaporizers used on Ohmeda Anesthesia Systems, including: Ohmeda Excel 110, 210, and 210SE; Ohmeda Modulus II and Modulus II Plus; Ohmeda Modulus SE; Ohmeda Modulus CD. Recall #Z-642-8.
CODE
Vaporizers serviced by Datex-Ohmeda from 7/11/97 to 4/14/98.
Subject Tec 6 Vaporizers have individual serial numbers which are
not sequential.
MANUFACTURER
Datex-Ohmeda (formerly Ohmeda Medical Systems Division), Madison,
Wisconsin.
RECALLED BY
Manufacturer, by letter on June 22, 1998. Firm-initiated recall
ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
6,207 vaporizers were distributed.
REASON
Gas leaks.
PRODUCT
a) 3M Modular Shoulder System, 14 mm X 125 mm stems, product
number 17464; b) 3M Modular Should System, 16 mm X 125 mm stems,
product number 17466, a metal stem designed to fit into a
surgically-prepared cavity in a humerus bone during shoulder
joint replacement surgery. The stems are designed for "press-fit
applications", but may also be cemented. Recall #Z-643/644-8
CODE
Lot numbers: a) 96 G 398; b) 96 G 348.
MANUFACTURER
3M Health Care, Rotherham, England.
RECALLED BY
3M Health Care, St. Paul, Minnesota, by letter dated June 16,
1998. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
a) 42 stems; b) 31 stems were distributed.
REASON
One stem with a 16 mm diameter was in a box labeled as 14 mm
diameter.
PRODUCT
Medex Injection Adapter, with Male Leur Lock, intended for use in
the administration of small volume fluids, using syringes.
Recall #Z-645-8.
CODE
Catalog #MX492, Lot #27L011175.
MANUFACTURER
Medex, Inc., Dublin, Ohio (packager).
RECALLED BY
Medex, Inc., Hillard, Ohio, by letter on June 26, 1998.
Firm-initiated recall ongoing.
DISTRIBUTION
Arkansas, Illinois, New Jersey, Pennsylvania, New York, Montana,
Connecticut, California, Maryland, Missouri, North Carolina,
Tennessee, Texas, Virginia, Iowa, Oregon, Minnesota.
QUANTITY
390 cases (100 units per case) were distributed.
REASON
The sterility of the device cannot be assured due to defective
packaging.
PRODUCT
Newport Infant/Adult "Breeze" Ventilator, designed for
respiratory support of infants, pediatrics, or adults:
a) Model E200D (domestic sales) Ventilator;
b) Model E200E (foreign sales) Ventilators.
Recall #Z-646/647-8.
CODE
The serial numbers of the 61 ventilators under recall are as
follows: 9704WD143; 9704WD147; 9709WD312; 9711WE396 through
9711WE415; 9712WE416; 9712WE418; 9712WE421 through 9712WE430;
9712WE432; 9712WE433; 9712WE435 through 9712WE437; 9712WE439
through 9712WE441; 9712WE443; 9712WE444; 9712WE446 through
9712WE449; 9712WE451; 9712WE452; 9712WE454; 9801WE001 through
9801WE003; 9801WE005 through 9801WE007; 9801WE023; 9801WE026;
9802WE032; 9802WE038.
MANUFACTURER
Mer-Mar, Inc., Hesperia, California (PC Board); Harris Central
Application, Palm Bay, Florida (chip).
RECALLED BY
Newport Medical Instruments, Inc., (NMI) Costa Mesa, California,
by letter on March 5, 1998. Firm-initiated field correction
ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
61 ventilators were distributed.
REASON
A bad IC (chip) causes inadvertent System Reset, continually
during power-up, or randomly during operation, wherein (1) the
ventilators cease operation (gas flow ceases and the exhalation
valve opens) for about 4 seconds, then restarts at values that
were set before the reset (except as follows), (2) the Sigh
Function defaults to off, (3) the Nebulizer defaults to off, (4)
the Inspiratory Minute Volume Alarm defaults to the 0-50 L. range
(an increase by a factor of 10 if originally set in the 1-5 L.
range), (5) Apnea Back-up Ventilation will likely commence for
foreign units (at 1.5 times the set rate, but with a minimum of
15 breaths/min and a maximum of 100 breaths/min with previously
set tidal volumes and pressures), and (6) the low pressure alarm
will likely reset, leading to nuisance alarms.
PRODUCT
Victoreen VeriDose Diodes, designed for use as radiation
detection devices encased within a polystyrene material:
a) Model 30-472 positive polarity VeriDose diodes;
b) Model 30-472-8000 negative polarity VeriDose Diodes.
Recall #Z-649/650-8.
CODE
All serial numbers manufactured since August 1997.
MANUFACTURER
Victoreen, Inc., Cleveland, Ohio.
RECALLED BY
Manufacturer, by telephone on May 7, 1998.
Firm-initiated recall ongoing.
DISTRIBUTION
Arizona, Kentucky, New York, Pennsylvania, Florida, Virginia,
Maryland, Michigan, Wisconsin, Ohio, Tennessee, Texas, Alabama,
Georgia, Mississippi.
QUANTITY
23 diodes were distributed.
REASON
The devices were manufactured with an inappropriate shielding
material (build-up cap). The diodes are intended for use at
energies of 5 to 11 MV, while the build-up caps are intended for
use in diodes for energies of 1 to 4 MV.
PRODUCT
Tracheal Intubation Fiberscope Instruction Manual:
a) Catalog No. 8080201, LF-P Instruction Manual;
b) Catalog No. 8080202, LF-2 Instruction Manual.
Recall #Z-667/668-8.
CODE
Incorrect Manual Labeling Identification Numbers:
Serial numbers of affected units:
a) 1700832 1700833 1700840 1810903 1700844 1700851
1700828 1700855 1700827 1700831 1810900 1700841
1700856 1700857
b) 2715299 2715184 2715296 2715277 2815337 2815313
2715179 2815381 2815406 2715281 2815419 2715173
2715170 2715235 2715242 2715258 2715265 2715304
2815420 2815422 2815423 2715260 2715266 2715269
2715270 2715172 2715214 2825457 2815317 2715252
2715233 2715234 2715245 2815325 2815442 2715202
2715232 2815412 2815438 2815371 2815414 2815348
2815353 2815417 2815360 2815355 2715272 2715291
2715186 2815413 2715271 2715221 2815408 2715308
2715294 2715192 2715209 2715286 2825468 2815362
2815370 2715162 2715178 2715185 2815350 2815364
2715241 2815349 2815384 2715158 2715163 2715204
2815345 2715177 2815380 2715227 2815319 2715276
2715279 2815389 2815398 2815396 2815418 2715297
2815339 2715169 2715231 2715191 2815378 2715180
2715287 2715273 2715290 2715152 2815410 2815309
2815344 2715298 2815436 2825326 2815376 2715292
2815328 2815315 2815405 2715217 2715155 2715164
2715278 2815386 2815363 2815428 2715189 2715244
2715239 2815414 2815359 2815335 2815427 2715267
2715284 2715240 2815349 2715167 2715183 2715230
2715237 2715274 2715157 2825454 2815352 2815377
2825466 2715205 2815415 2715254 2715226 2825452
2715302 2715196 2715181 2715176 2815340 2815356
2815327 2715171 2715215 2815387 2815404 2715289
2715301 2715238 2715187 2815416 2815388 2815367
2715282 2715293 2815409 2715280 2815347 2715307
2715211 2815382 2715168 2815383 2815401 2715199
2715198 2715220 2815421 2715300 2715288.
MANUFACTURER
Olympus Optical Company, Ltd., Hinode Plant, Tokyo, Japan and
Olympus Opto-Electronics Company, Ltd., Fufushima-ken, Japan.
RECALLED BY
Olympus American, Inc., Melville, New York (distributor), by
letter on or about June 3, 1998. Firm-initiated recall
ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
a) 27 units; b) 287 units were distributed were distributed.
REASON
The cover pages for the instruction manuals were for the correct
intubation fiberscope, but the rest of the instructions were for
the other one.
PRODUCT
Model Tru-Pulse C02 Laser System, used in general surgery.
Recall #Z-669-8.
CODE
Model True-Pulse C02.
MANUFACTURER
Tissue Medical Lasers, Inc., Albuquerque, New Mexico.
RECALLED BY
Manufacturer. FDA approved the firm's corrective action plan on
July 1, 1998. Firm-initiated field correction ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
275 units were distributed.
REASON
Noncompliance with performance standards for laser products in
that the operators manual lacked calibration procedures.
PRODUCT
I-125 Seeds, (Iodine-125) Therapeutic for Interstitial
Brachytherapy, Model 6711, used as radioactive implants for the
treatment of prostate cancer. Recall #Z-670-8.
CODE
Lot W71536.
MANUFACTURER
Medi-Physics, Inc., Amersham Healthcare, Arlington Heights,
Illinois.
RECALLED BY
Manufacturer, by telephone on September 23, 1997. Firm-initiated
recall complete.
DISTRIBUTION
Georgia and Florida.
QUANTITY
26 seeds were distributed.
REASON
The seeds were mislabeled with an incorrect apparent
radioactivity range.
PRODUCT
I-125 Seeds, (Iodine-125) Therapeutic for Interstitial
Brachytherapy, Model 6711, used as radioactive implants for the
treatment of prostate cancer. Recall #Z-671-8.
CODE
Lot W72531.
MANUFACTURER
Medi-Physics, Inc., Amershlam Healthcare, Arlington Heights,
Illinois.
RECALLED BY
Manufacturer, by telephone on January 8, 1998. Firm-initiated
recall complete.
DISTRIBUTION
Washington state, California, New York.
QUANTITY
161 seeds were distributed.
REASON
The seeds were mislabeled with an incorrect apparent
radioactivity range.
PRODUCT
Disposable Manual Resuscitator, Part #DMR2, used to deliver
supplemental oxygen. Recall #Z-672-8.
CODE
None. Units shipped between January 28, 1998 and April 22,
1998.
MANUFACTURER
Nellcor Mexico, Tijuana, Mexico.
RECALLED BY
Nellcor Puritan Bennett, Carlsbad, California, by letter on May
22, 1998. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and Italy.
QUANTITY
67,490 units were distributed.
REASON
The duckbill valve can become unseated, allowing rebreathing
and/or insufficient inspiratory air pressure; and (2) the
pressure relief valve ball can rust, causing it to become
inoperative.
END OF ENFORCEMENT REPORT FOR JULY 15, 1998. BLANK PAGES MAY FOLLOW.