FDA

Enforcement Report

The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

  May 27, 1998                                         98-21
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II =========
  _______________
  PRODUCT      Post Type Cylinder Valves, for medical gas
               cylinders.  Recall #D-134-8.
  CODE         All lot codes.
  MANUFACTURER Ceodeux, Inc., Ultrapure Equipment Technology
               S.A., Lintgen/Luxembourg.
  RECALLED BY  Rotarex, Inc., North America, Mt. Pleasant,
               Pennsylvania, by letter dated March 25, 1998. 
               Firm-initiated field correction ongoing.
  DISTRIBUTION Massachusetts, Georgia, California,
               Pennsylvania, Maryland, North Carolina,
               Kansas, Canada.
  QUANTITY     Approximately 125,000 units were manufactured
               and of that approximately 43,750 remained on
               market at time of recall initiation.
  REASON       High velocity valve stem ejection.
  
  _______________
  PRODUCT      Sterile Vancomycin Hydrochloride, USP,
               lyophilized, 10g/100mL vial, pharmacy bulk
               package, not for direct infusion, Rx, 
               lyophilized powder for preparing intravenous
               infusions for the treatment of serious or
               severe infections caused by susceptible
               strains of methicillin-resistant
               staphylococci.  NDC #0469-3140-61.
               Recall #D-144-8.
    CODE           Lot numbers 170438 and 170449. 
MANUFACTURER   Fujisawa USA, Inc., Melrose Park, Illinois.
RECALLED BY    Fujisawa USA, Inc., Deerfield, Illinois, by
               letter dated April 30, 1998.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       3,455 vials were distributed; firm estimated
               that 25 percent remained on market at time of
               recall initiation.
REASON         Incorrect volumes - Vials are labeled with
               graduation marks that do not indicate an
               accurate volume.
  
  _______________
PRODUCT        Propanolol HCL Capsules, Extended Release, 
               80 mg, in units of 30, indicated in the
               management of hypertension.  
               NDC #54569-3097-0.  Recall #D-146-8.
CODE           7140089, 7164046, 7176062, 7293083, 7311102.
MANUFACTURER   Inwood Laboratories, Inc., Inwood, New York
               (responsible firm).
RECALLED BY    Allscripts, Inc., Libertyville, Illinois
               (repackager), by letter dated April 6, 1998. 
               Firm-initiated recall ongoing.
DISTRIBUTION   New Hampshire, Iowa, Florida, New Mexico,
               Illinois.
QUANTITY       103 bottles were distributed; firm estimated
               that little, if any, product remained on
               market at time of recall initiation.
REASON         Dissolution failure.
  
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
  _______________
  PRODUCT      Ranitidine Tablets, USP (Ranitidine
                 Hydrochloride), 150 mg, in 500 tablet bottles,
                 Rx for the short-term treatment of active
                 duodenal ulcer.  NDC #55953-544-70.
               Recall #D-145-8.
  CODE         Lot #108969C.
  MANUFACTURER Novopharm, Ltd, Scarborough, Ontario, Canada.
  RECALLED BY  Novopharm USA, Inc., Schaumburg, Illinois, by
                 telephone on May 5, 1998.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION Maine, Florida, Louisiana, Pennsylvania,
                 Wisconsin, Missouri, Alabama, South Dakota,
                 Georgia, Ohio, Tennessee, Connecticut,
                 Minnesota.
  QUANTITY     1,638 bottles were distributed; firm estimated
                 that 80 percent of the product remained on
                 market at time of recall initiation.
  REASON       Incorrect bottle desiccant.
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-549-8.
  CODE         Unit #24196-6241.
  MANUFACTURER Blood Systems, Inc., Rapid City, South Dakota.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 telephone on July 31, 1997, and by letter
                 dated August 27, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION Illinois.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 reported travel to an area designated as
                 endemic for malaria.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c) Recovered
                 Plasma, Frozen.  Recall #B-795/797-8.
  CODE         Unit #28GK19453.
  MANUFACTURER American Red Cross Blood Services, Savannah,
                 Georgia.
  RECALLED BY  Manufacturer, by letters dated October 10 and
                 29, 1997, and November 11, 1997.  Firm-initiated
recall complete.
  DISTRIBUTION Illinois, South Carolina, California.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 with a history of cancer.
  
  _______________
  PRODUCT      Source Plasma.  Recall #B-846-8.
  CODE         Unit #3448W.
  MANUFACTURER Concho Biologics, Inc., San Angelo, Texas.
  RECALLED BY  Manufacturer, by telephone and fax on August
                 25, 1997.  Firm-initiated recall complete.
  DISTRIBUTION North Carolina.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 previously tested reactive for syphilis within
                 one year of donation.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-991-8
  CODE         Unit GW29790
  MANUFACTURER Metropolitan Washington Blood Banks, Bethesda,
                 Maryland.
  RECALLED BY  Manufacturer, by letter dated May 9, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested repeatedly reactive for
                 the antibody to the Hepatitis B core antigen
                 (anti-HBc).
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-992-8.
  CODE         Unit GW27208
  MANUFACTURER Metropolitan Washington Blood Banks, Bethesda,
                 Maryland.
  RECALLED BY  Manufacturer, by telephone and letter on March
                 1, 1997.  Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON            Blood product tested non-reactive for the
                      antibody to the Hepatitis B core antigen
                      (anti-HBc), but was collected from a donor
who
                      was previously repeatedly reactive for
                      anti-HBc on two separate donations.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1009-8.
  CODE         Unit #3878789.
  MANUFACTURER Camp Memorial Blood Center, Fort Knox,
                 Kentucky.
  RECALLED BY  Manufacturer, by fax dated August 7 or 8,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Georgia and Missouri.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 received either a tattoo, body piercing,
                 acupuncture, or accidental needle stick,
                 within twelve months of donation.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c) Recovered
                 Plasma.  Recall #B-1078/1080-9.
  CODE         Unit numbers:  a) 18Y86733 and 18W87490
               b) 18Y86733; c) 18Y86733 and 18W87490.
  MANUFACTURER American Red Cross Great Lakes Region, 
                 Lansing, Michigan.
  RECALLED BY  Manufacturer, by letter either on February 24
                 or 27, 1998, by telephone on February 27,
                 1998, and by second letter on March 10 or 19,
                 1998.  Firm-initiated recall complete.
  DISTRIBUTION Michigan and California.
  QUANTITY     a) 2 units; b) 1 unit; c) 2 units were
                 distributed.
  REASON       Blood products were collected from a donor
                 whose spouse had active hepatitis C.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c) Recovered
                 Plasma, frozen.  Recall #B-1081/1083-8.
  CODE         Unit #1017534.
  MANUFACTURER Community Blood Center, Dayton, Ohio.
  RECALLED BY  Manufacturer, by letter dated August 29, 1997,
                 or September 9, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION Indiana, Ohio, California.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 with a history of jaundice.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1085-8.
  CODE         Unit #GW32455.
  MANUFACTURER Metropolitan Washington Blood Banks, Inc.,
                 Bethesda, Maryland.
  RECALLED BY  Manufacturer, by telephone on August 12, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Maryland.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested positive for syphilis.
  
  _______________
  PRODUCT      Platelets, Pheresis.  Recall #B-1086-8.
  CODE         Unit #KZ24929.
  MANUFACTURER Virginia Blood Services, Richmond, Virginia.
  RECALLED BY  Metropolitan Washington Blood Banks, Inc.,
                 Bethesda, Maryland, by telephone and fax on
                 December 8, 1995.  Firm-initiated recall
                 complete.
  DISTRIBUTION North Carolina.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was labeled with an extended
                 expiration date.
  
  _______________
  PRODUCT      a) Platelets, Pooled; b) Fresh Frozen Plasma. 
                 Recall #B-1095/1096-8.
  CODE         Unit numbers:  a) P62602; b) 9137888.
  MANUFACTURER Carter Blood Center, Fort Worth, Texas.
  RECALLED BY  Manufacturer, by telephone on March 12, 1998,
                 and by letter dated March 13, 1998. 
Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products corresponded to a unit of
                 clotted red blood cells.
  
  _______________
  PRODUCT      a) Platelets; b) Fresh Frozen Plasma.  
               Recall #B-1097/1098-8.
  CODE         Unit #9138815.
  MANUFACTURER Carter Blood Center, Fort Worth, Texas.
  RECALLED BY  Manufacturer, by telephone on March 14, 1998,
                 and by letter dated April 16, 1998. 
Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products corresponded to a unit of
                 clotted red blood cells.
  
  _______________
  PRODUCT      a) Platelets, Pooled; b) Fresh Frozen Plasma.
               Recall #B-1099/1100-8.
  CODE         Unit numbers:  a) P62626; b) 9232101.
  MANUFACTURER Carter Blood Center, Fort Worth, Texas.
  RECALLED BY  Manufacturer, by telephone on March 12, 1998,
                 and by letter dated March 13, 1998. 
Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products corresponded to a unit of
                 clotted red blood cells.
  
  _______________
  PRODUCT      Platelets, Pooled.  Recall #B-1101-8.
  CODE         Unit #P62698.
  MANUFACTURER Carter Blood Center, Fort Worth, Texas.
  RECALLED BY  Manufacturer, by telephone on March 11, 1998,
                 and by letter dated March 16, 1998. 
Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     1 unit was distributed.
  REASON       Blood product corresponded to a unit of
                 clotted red blood cells.
  
  
  _______________
  PRODUCT      Platelet Pheresis.  Recall #B-1103-8.
  CODE         Unit #21 P76619.
  MANUFACTURER American Red Cross, Portland, Oregon.
  RECALLED BY  Manufacturer, by letter dated April 3, 1998. 
                 Firm-initiated recall complete.
  DISTRIBUTION Oregon.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 received a tattoo within 12 months of
                 donation.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1105-8.
  CODE         Unit numbers:  21GC20021 and 21GL30711.
  MANUFACTURER American Red Cross, Portland, Oregon.
  RECALLED BY  Manufacturer, by letter dated March 18, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Oregon and California.
  QUANTITY     2 units were distributed.
  REASON       Blood products were collected from a donor
                 with a history of cancer.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-1106/1107-8.
  CODE         Unit #21365-8659.
  MANUFACTURER Blood Systems, Inc., United Blood Services,
                 Fargo, North Dakota.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 letters dated October 6, 1997, and April 20,
                 1998.  Firm-initiated recall complete.
  DISTRIBUTION Illinois and Switzerland.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 with a history of hepatitis.
  
  _______________
  PRODUCT      Source Plasma.  Recall #B-1108-8.
  CODE         Unit numbers:  33-ZBP950-0, 33-ZBQ284-0,
                 33-ZBQ417-0, 33ZBQ715-0.
  MANUFACTURER Centeon Bio-Services, Inc., West Allis,
                 Wisconsin.
  RECALLED BY  Centeon Bio-Services, Inc., Knoxville,
                 Tennessee, by an electronic message dated
                 November 11, 1996.  Firm-initiated recall
                 complete.
  DISTRIBUTION Germany.
  QUANTITY     4 units were distributed.
  REASON       Blood products were collected from an
                 unsuitable donor.
  
  
  _______________
  PRODUCT      Source Plasma.  Recall #B-1109-8.
  CODE         Unit numbers:  33-ZB7452-B, 33-ZB7515-B,
                 33-ZB7507-B, 33-ZB7563-B
  MANUFACTURER Centeon Bio-Services, Inc., West Allis,
                 Wisconsin.
  RECALLED BY  Centeon Bio-Services, Inc., Knoxville,
                 Tennessee, by an electronic message dated
                 November 20, 1996.  Firm-initiated recall
                 complete.
  DISTRIBUTION Germany.
  QUANTITY     4 units were distributed.
  REASON       Blood products were collected from unsuitable
                 donors.
  
  
  _______________
  PRODUCT      Recovered Plasma.  Recall #B-1111-8.
  CODE         Unit #GW06654.
  MANUFACTURER Metropolitan Washington Blood Banks, Inc.,
                 Bethesda, Maryland.
  RECALLED BY  Manufacturer, by telephone on December 27,
                 1994.  Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested repeatedly reactive for
                 the antibody to the hepatitis C virus encoded
                 antigen (anti-HCV).
  
  _______________
  PRODUCT      Source Plasma.  Recall #B-1112-8.
  CODE         Unit #DR071359.
  MANUFACTURER Sera-Tec Biologicals Limited Partnership, Del
                 Rio, Texas.
  RECALLED BY  Sera-Tec Biologicals Limited Partnership,
                 Harrisburg, Pennsylvania, by telefax on
                 December 3, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION Indiana.
  QUANTITY     1 unit was distributed.
  REASON       Blood product had an elevated ALT.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1113-8.
  CODE         Unit #53GT06124.
  MANUFACTURER American Red Cross Blood Services, Baltimore,
                 Maryland.
  RECALLED BY  Manufacturer, by letter dated February 20,
                 1998.  Firm-initiated recall complete.
  DISTRIBUTION Maryland.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested positive for anti-Rho(D)
                 antibodies.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets.  
               Recall #B-1114/1115-8.
  CODE         Unit numbers:  a) KH67776, KH68200, KH69840,
                 KH68861, KH70977, KH71003, KH73915, KH75739
               b) KH67776, KH69840, KH68861, KH70977,
                 KH71003, KH73915.
  MANUFACTURER Virginia Blood Services, Richmond, Virginia.
  RECALLED BY  Manufacturer, by letter dated April 10, 1998.
                 Firm-initiated recall complete.
  DISTRIBUTION Maryland and Virginia.
  QUANTITY     a) 8 units; b) 6 units were distributed.
  REASON       Blood products, collected from donors who
                 traveled to areas considered endemic for
                 malaria.
  
  _______________
  PRODUCT      Platelets, Pheresis.  Recall #B-1116-8.
  CODE         Unit #0310189A.
  MANUFACTURER Inland Northwest Blood Center, Spokane,
                 Washington.
  RECALLED BY  Manufacturer, by telephone on December 19,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Washington state.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was labeled with an extended
                 expiration date.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets.  
               Recall #B-1117/1118-8.
  CODE         Unit #GW39677.
  MANUFACTURER Metropolitan Washington Blood Bank, Bethesda,
                 Maryland.
  RECALLED BY  Manufacturer, by letter dated April 27, 1998. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Virginia.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor who
                 reported travel to an area designated as
                 endemic for malaria.
  
  _______________
  PRODUCT      Red Blood Cells. Recall #B-1119-8.
  CODE         Unit #21FM34832.
  MANUFACTURER American Red Cross, Portland, Oregon.
  RECALLED BY  Manufacturer, by letter dated March 10, 1998. 
                 Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor
                 taking the drug Proscar.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c) Recovered
                 Plasma.  Recall #B-1120/1122-8.
  CODE         Unit #42S53953.
  MANUFACTURER American Red Cross Blood Services, Cleveland,
                 Ohio.
  RECALLED BY  Manufacturer, by letters dated February 17 &
                 19, 1997.  Firm-initiated recall complete.
  DISTRIBUTION Ohio and California.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 taking the drug Lupron.
  
  _______________
  PRODUCT      Fresh Frozen Plasma.  Recall #B-1123-8.
  CODE         Unit #17KX25737.
  MANUFACTURER American Red Cross Blood Services, St Paul,
                 Minnesota.
  RECALLED BY  Manufacturer, by telephone on September 30,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Minnesota.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor with
                 a history of cancer.
  
  _______________
  PRODUCT      Recovered Plasma.  Recall #B-1126-8.
  CODE         Unit #17FC88836.
  MANUFACTURER American Red Cross Blood Services, St. Paul,
                 Minnesota.
  RECALLED BY  Manufacturer, by fax on November 6, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor with
                 a history of cancer.
  
  _______________
  PRODUCT      Recovered Plasma.  Recall #B-1127-8.
  CODE         Unit numbers:  17KP40578 and 17GL50917
  MANUFACTURER American Red Cross Blood Services, North
                 Central Blood Services,               St. Paul,
Minnesota.
  RECALLED BY  Manufacturer, by fax on February 18, 1998. 
                 Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     2 units were distributed.
  REASON       Blood products were collected from a donor
                 with a history of cancer. 
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1138-8.
  CODE         Unit #53KT47365.
  MANUFACTURER American Red Cross Blood Services, Baltimore,
                 Maryland.
  RECALLED BY  Manufacturer, by telephone on November 11,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION District of Columbia.
  QUANTITY     1 unit was distributed.
  REASON       Blood product corresponded to a unit of plasma
                 that was hemolyzed.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma. 
                 Recall #B-1140/1141-8.
  CODE         Unit #24021-4622.
  MANUFACTURER Blood Systems, Inc., United Blood Services,
                 Rapid City, South Dakota.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 letters dated December 4, 1997, and April 20,
                 1998.  Firm-initiated recall complete.
  DISTRIBUTION Illinois and Switzerland.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor who
                 may have received an injection of Hepatitis B
                 Immune Globulin within 12 months of donation.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-1142/1143-8.
  CODE         Unit numbers:  24GS12055, 24GS18888
  MANUFACTURER American Red Cross Blood Services, Louisville,
                 Kentucky.
  RECALLED BY  Manufacturer, by letter dated March 18 or 19,
                 1998, and March 16, 1998.  Firm-initiated
                 recall complete.
  DISTRIBUTION Indiana, Ohio, California.
  QUANTITY     2 units of each component were distributed.
  REASON       Blood products tested negative for the
                 antibody to the human immunodeficiency virus
                 type 1 (HIV-1), but were collected from donors
                 who previously tested repeatedly reactive for
                 the antibody, and was subject to additional
                 testing.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1144-8.
  CODE         Unit #36LC03141.
  MANUFACTURER American Red Cross Blood Services, Columbia,
                 South Carolina.
  RECALLED BY  Manufacturer, by letter dated September 4,
                 1996.  Firm-initiated recall complete.
  DISTRIBUTION South Carolina.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was prepared from a unit of
                 whole blood with an extended collection time.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-1146/1147-8.
  CODE         Unit #42FH66597.
  MANUFACTURER American Red Cross Blood Services, Cleveland,
                 Ohio.
  RECALLED BY  Manufacturer, by telephone on September 23 and
                 24, 1997, and by letter on September 24, 1998. 
                 Firm-initiated recall complete.
  DISTRIBUTION Ohio and California.
  QUANTITY     1 unit was distributed.
  REASON       Blood products were incorrectly tested for the
                 hepatitis B surface antigen (HbsAg).
  
  _______________
  PRODUCT      Recovered Plasma.  Recall #B-1148-8.
  CODE         Unit #GW43585.
  MANUFACTURER Metropolitan Washington Blood Bank, Bethesda,
                 Maryland.
  RECALLED BY  Manufacturer, by fax on April 14, 1998,
                 followed by letter on April 27, 1998. 
Firm-initiated recall complete.
  DISTRIBUTION Switzerland.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested repeatedly reactive for
                 the antibody to the human immunodeficiency
                 virus types 1 and 2 (anti-HIV-1/2).
  
  _______________
  PRODUCT      Source Plasma.  Recall #B-1150-8.
  CODE         Unit #XZ94131, XZ94132, XZ94381.
  MANUFACTURER Community Bio-Resources, Inc., Eau Claire,
                 Wisconsin.
  RECALLED BY  Community Bio-Resources, Inc., Hoover,
                 Alabama, by letter dated September 18, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Michigan.
  QUANTITY     3 units were distributed.
  REASON       Blood products, which were not tested for
                 syphilis.
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1023-8.
  CODE         Unit #3872515.
  MANUFACTURER Camp Memorial Blood Center, Fort Knox,
                 Kentucky.
  RECALLED BY  Manufacturer, by fax dated September 12 or 16,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Texas and California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 exceeded the donation acceptance criteria for
                 body temperature.
  
  _______________
  PRODUCT      Autologous Red Blood Cells.  Recall #B-1084-8.
  CODE         Unit #25-17833.
  MANUFACTURER Michigan Community Blood Centers, Grand
                 Rapids, Michigan.
  RECALLED BY  Manufacturer, by telephone on August 12, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Michigan.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested positive for the antibody
                 to the hepatitis C virus encoded antigen
                 (anti-HCV), was distributed without a
                 biohazard label or physician approval.
  
  _______________
  PRODUCT      a) Platelets; b) Platelets, Pheresis.
               Recall #B-1087/1088-8.
  CODE         Unit numbers:  a) 10530-9056, 10531-1112,
                 10530-9096, 10530-7173, 10531-0485,
                 10531-0483, 10531-0106, 10531-1102
               b) 10530-0108, 10530-8638, 10530-0125,
                 10530-0128, 10530-6151, 10530-6157.
  MANUFACTURER United Blood Services, Scottsdale, Arizona.
  RECALLED BY  Blood Systems, Inc., Scottsdale, Arizona, by
                 letter dated July 24, 1997.  Firm-initiated
                 recall complete.
  DISTRIBUTION Arizona.
  QUANTITY     a) 8 units; b) 6 units were distributed.
  REASON       Blood products were stored at an unacceptable
                 temperature.
  
  _______________
  PRODUCT      Fresh Frozen Plasma.  Recall #B-1102-8.
  CODE         Unit #9138817.
  MANUFACTURER Carter Blood Center, Fort Worth, Texas.
  RECALLED BY  Manufacturer, by telephone on March 31, 1998,
                 and by letter dated April 2, 1998. 
Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a unit of
                 whole blood with an extended collection time.
  
  
 RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS
II =========
  _______________
  PRODUCT      Isolette, Model C2HS (C2000), incubator used
               for critically ill newborn premature babies. 
               Recall #Z-570-8.
  CODE         All product marketed since November 1996.
  MANUFACTURER Hill-Rom Air-Shields, Hatboro, Pennsylvania.
  RECALLED BY  Manufacturer, by letter dated May 18, 1998. 
               Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and international.
  QUANTITY     Approximately 2000 units were distributed.
  REASON       The controller and humidity module are subject
               to temperature fluctuations, humidity
               departures from set points and air flow probe
               failures.
  
  _______________
  PRODUCT      CORFLO - CuBBy LPDG Low Profile Gastrostomy
               Device; an Rx sterile single patient use
               device consisting of an inner balloon that
               rests against the patient's stomach wall and
               an outer portion that rests on the patient's
               skin, which provides access to the patient's
               stomach for enteral feedings and medications:
               a) Catalog No. 31-1820, 18FR, 2.0 cm length;
               b) Catalog No. 31-2020, 20FR, 2.0 cm length.  
               Recall #Z-578/579-8.
  CODE         Lot numbers:  a) A01LK and A01LK-1;
               b) A01IB.
  MANUFACTURER Manufacturing & Research, Inc. (MRI), Tucson,
               Arizona.
  RECALLED BY  Corpak MedSystems, Wheeling, Illinois, by
               letter dated May 1, 1998.  Firm-initiated
               recall ongoing.
  DISTRIBUTION Pennsylvania, Florida, California,
               Connecticut, Missouri, Texas, Maryland, South
               Carolina, New York, Oklahoma, Michigan,
               District of Columbia, North Carolina,
               Tennessee and Kentucky.
  QUANTITY     104 units of 31-1820 and 25 units of 31-2020
               were distributed between November 1997 through
               April 1998, firm estimated that 20% remained
               on the market at time of recall initiation.
  REASON       There may be separation or partial separation
               of the tube from the dome of the device.
  
  
  RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
  _______________
  PRODUCT      Kendall Sher-I-Bronch Left-Sided Endobronchial
               Tube a) Product Code: 5-16039 (39Fr); b) Product
               Code: 5-16037 (37Fr)  , indicated for
               use in thoracic surgery, bronchospirometry,
               for the administration of endobronchial
               anesthesia and other uses commonly associated
               with endobronchial tubes.  
               Recall #Z-573/574-8.
  CODE         Lot Numbers: a) 063534; b) 063532.
  MANUFACTURER Kendall Sheridan, Argyle, New York.
  RECALLED BY  Kendall Healthcare Products Company,
               Mansfield, Massachusetts, by letter between
               March 23 and 25, 1998.  Firm-initiated recall
               ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     a) 306 units; b) 312 units were distributed.       
             
  REASON       Unit pouch is mislabeled with incorrect French
               size.
  
  _______________
  PRODUCT      Suture Retrievers:  a) BTS Mini-Laparotomy MMK
               Set, Model No. UM250 (Part  No. 120089) 
               b) Laparotomy Bladder Neck Suspension Kit,   
               Model No. MW-100 
               c) BTS Urethropexy Suture Placement Set,     
               Model No. UM-300, (Part No. 120148).  
               Recall #Z-575/577-8.
  CODE         Lot numbers:  a) 9ML80206-04;
               b) 9ML7015-01, 9ML70814-01, 9ML71021-04,  
               9SL71023-7,  9ML71023-01;
               c) 9ML80206-01, 9ML71201-03.
  MANUFACTURER Louisville Laboratories, Inc., Louisville,
               Kentucky.
  RECALLED BY  Medworks Corporation, Louisville, Kentucky, by
               letter dated April 16, 1998.  Firm-initiated
               recall ongoing.
  DISTRIBUTION California, Colorado, Florida, Georgia,
               Illinois, Kentucky, Massachusetts, Maryland,
               Missouri, New Jersey, Nevada, New York
               Oklahoma, Pennsylvania, Texas, Virginia,
               Washington state.
  QUANTITY     a) 52 devices; b) 43 devices; c) 97 devices
               were distributed.
  REASON       That the plastic handle of the suture
               retriever may crack at the point where the
               stainless steel needle attaches to the plastic
               handle.
  
  
  RECALLS AND FIELD CORRECTIONS:  VETERINARY PRODUCTS -- CLASS II
===========
  _______________
  PRODUCT      Dexamethasone sodium phosphate injection - 4
                 mg/ml (used as an anti-inflammatory sterile
                 intravenous injection).  Product is packaged
                 with two labels:
               Steris Laboratories, Inc., Phoenix, Arizona-
                 Dexamethasone Sodium Phosphate Injection Vetus
                 Animal Health, Dexaject SP., distributed by
                 Burns Veterinary Supply and manufactured by
                 Steris Laboratories, Inc.  Recall #V-022-8.
  CODE         Lot numbers:  96G832, 96K080, 96L460, 96L830,      
                                                
                 97A340, 97B810, 97E460, 97H260, 97H570,
                 97H571, 97H700, 97H701, 97L230, 97L231.
  MANUFACTURER Steris Laboratories, Inc., Phoenix, Arizona.
  RECALLED BY  Manufacturer, by letter dated April 22, 1998. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and Puerto Rico (Steris label);
                 Texas (Vetus label).
  QUANTITY     QUANTITY DISTRIBUTED      QUANTITY DISTRIBUTED
                 Lot #     STERIS LABEL           VETUS LABEL
                 96G832    12036                     2996
                 96K080    17820                     0
                 96L460    14870                     3022
                 96L830    10895                     3021
                 97A340    16279                     3019
                 97B810    18324                     0
                 97E460    18012                     2952
                 97H260    17724                     0
                 97H570    8820                      0
                 97H571    1416                      0
                 97H700    5328                      3023
                 97H701    8880                      0
                 97L230    3588                      0
                 97L231    4734                      0.
  REASON       Stability data does not support labeled
                 expiration dates.
  
  
  MEDICAL DEVICE SAFETY ALERTS:  ==============================
  _______________
  UPDATE       Recall #Z-561-8, Adult BagEasy Disposable
                 Manual Resuscitators (All Types and Models),
                 Lot Nos. A960521 through A971201 and B960521
                 through B971201 (Respironics, Murrysville,
                 PA), which appeared in the May 13, 1998
                 Enforcement Report, has been reclassified as a
                 voluntary Safety Alert and assigned Safety
                 Alert #N-005-8.
  
  
  END OF ENFORCEMENT REPORT FOR MAY 27, 1998.

                                   ####


[FDA HOME PAGE]