The FDA Enforcement Report is published weekly by the Food and Drug Administration, U.S. Public Health Service, Department of Health and Human Services. It contains information on actions taken in connection with agency regulatory activities.

  
  April 8, 1998                                           98-14
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
  _______________
  PRODUCT      Wang Derm brand Salted Scomber Fish, also
                 called Pla Too in Thai), vacuum packed in
                 flexible plastic bags, approximately 8 ounces
                 per bag.  Recall #F-192-8.
  CODE         None.
  MANUFACTURER Wang Derm Exporter L.P., Bangkok, Thailand.
  RECALLED BY  Vasinee Food Corporation, Brooklyn, New York
                 (importer), by letter dated November 25, 1997,
                 and by fax on December 17, 1997.  Completed
                 recall resulted from sample analysis and
                 follow-up by the New York State Department of
                 Agriculture and Markets.
  DISTRIBUTION New York, Massachusetts, Connecticut,
                 Maryland, New Jersey, Illinois, Rhode Island,
                 Pennsylvania, Tennessee.
  QUANTITY     102 cases (36 bags per case) were distributed.
  REASON       The product, uneviscerated dried fish,  has
                 the potential to support the growth of and
                 toxin production by Clostridium botulinum.
  
    _______________
  PRODUCT      Rich's Carrot Cake, 8" Uniced, Round Layer
                 Cake.  Recall #F-194-8.
  CODE         All lots currently within expiration date. 
                 Products coded 7152 thru 8056 are subject to
                 recall.
  MANUFACTURER Country Oven Bakery, Bowling Green, Kentucky.
  RECALLED BY  Rich Products Corporation, Buffalo, New York,
                 by memorandum and letter dated March 2, 1998. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and United Arab Emirates.
  QUANTITY     4,550 cases (24 cakes per case) were
                 distributed.
  REASON       Product contains undeclared eggs.
  
  _______________
  PRODUCT      a) Dominick's Bakery Fudge Iced Brownies,
               whole, chocolate brownie with icing in a
                 square foil pan with a clear plastic lid,
                 approximately 18 ounces and 7" square
               b) Brownie Slices, square iced chocolate
                 brownie slices in a hinged clear plastic dome
                 package, 2 slice package and 6 slice package. 
                 Recall #F-290/291-8.
  CODE         All product with Sell By dates prior to Mar
                 22, 98.
  MANUFACTURER Dominick's Commissary, Northlake, Illinois.
  RECALLED BY  Dominick's Finer Foods, Inc., Northlake,
                 Illinois, by telephone on March 14, 1998,
                 followed by E-mail.  Firm-initiated recall
                 complete.
  DISTRIBUTION Illinois and Indiana.
  QUANTITY     2,521 packages of Fudge Iced Brownies, 14,287 
                 packages of 2 Brownie Slices, and 263 packages
                 of 6 Brownie Slices were distributed and sold. 
                 Firm estimates none of the products remain  on
                 the market.
  REASON       Products contain undeclared whole eggs.
  
  _______________
  PRODUCT      Fire Roasted Vegetable with Roasted Garlic and
                 Sundried Tomato Gourmet Style Gardenburger
                 Veggie Patties.   Recall #F-339-8.
  CODE         The individual boxes contain the following
                 codes:  012099, 020999, 021099, 022599,
                 022699, 030699, 021199, 021299, 021399.
  MANUFACTURER Gardenburger, Inc., Portland, Oregon.
  RECALLED BY  Manufacturer, by telephone beginning on March
                 17, 1998, and by press release on March 18,
                 1998.  Additional telephone calls, letters and
                 a press release were issued on April 2, 1998,
                 extending the recall to additional codes. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     2,287 cases (12 boxes per case) were
                 distributed.
  REASON       Product contains undeclared egg whites.
  
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
  _______________
  PRODUCT      DiaB Extra Comprehensive Wound Nutrition
                 Tablet Dietary Supplement, in bottles of 60. 
                 Recall #F-193-8.
  CODE         Lot Number 740410, no expiration date.
  MANUFACTURER Alpine Health Industries, Orem, Utah.
  RECALLED BY  Carrington Laboratories, Inc., Irving, Texas,
                 by letter dated July 15, 1998.  Firm-initiated
                 recall complete.
  DISTRIBUTION Nationwide and Canada.
  QUANTITY     1,429 bottles were distributed.  Firm
                 estimates none remains on the market.
  REASON       Product contains excessive levels of lead.
  
  _______________
  PRODUCT      Popcorn Tins that contain cheese, caramel, and 
                 regular popcorn with Christmas, NFL or
                 miscellaneous decorations in 3 1/2 gallon (38
                 oz.) or 5 1/2 gallon (60 ounce tins).
               Recall #F-289-8.
  CODE         All product which fails to declared FD&C

                 Yellow No. 5 & 6.
  MANUFACTURER Golden Harvest Products, Inc., Ontario,
                 California, and Houston Harvest Gift Products,
                 LLC, Overland Park, Kansas.
  RECALLED BY  Houston Harvest Gift Products, LLC, Overland
                 Park, Kansas, by fax letter dated December 12,
                 1997, followed by certified letter dated
                 February 26, 1998.  Firm-initiated recall
                 ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     7,174,767 38-ounce tins and 400,872 60-ounce
                 tins were distributed.
  REASON       Product contains undeclared FD&C Yellow No. 5
                 & 6.
  
  _______________
  PRODUCT      Caffeine Free Coca Cola Classic in 12 ounce
                 cans packed in 24-packs.  Recall #F-323-8.
  CODE         "SEP 21 98 HOE 1330" through "SEP 21 98 HOE
                 1342".
  MANUFACTURER Coca-Cola Bottling Company Indianapolis, Inc., 
                 Speedway, Indiana.
  RECALLED BY  Manufacturer, by press releases on January 15
                 and 16, 1998, and by visits on January 9 and
                 16, 1998.  Firm-initiated recall complete.
  DISTRIBUTION Indiana.
  QUANTITY     Approximately 281 cases (24 cans per case)
                 were distributed.
  REASON       Product is contaminated with ammonia, which
                 may cause it to be injurious to health.
  
  _______________
  PRODUCT      Basha Foods brand Hommus packaged in plastic
                 containers, a refrigerated dip-like product
                 made primarily of chick peas:
               a) Hommus Original in 8 ounce, 16 ounce and 5
                 pound containers;
               b) Hommus Vegetable, in 8 ounce containers
               c) Hommus South-of-the-Border, in 8 ounce
                 containers
               d) Hommus Roasted Red Pepper, in 8 ounce
                 containers
               e) Hommus Scallion, in 8 ounce containers
               f) Hommus Garlic, in 8 ounce containers
               g) Hommus Dill, in 8 ounce containers
               h) Hommus Zesty, in 8 ounce containers.
               Recall #F-324/331-8.
  CODE         All lots.
  MANUFACTURER Basha International Foods, Inc., Hamtramck,
                 Michigan.
  RECALLED BY  Manufacturer, by visit beginning on March 6,
                 1998.  Firm-initiated recall complete.
  DISTRIBUTION Colorado, Florida, Illinois, Michigan.
  QUANTITY     Approximately 32,000 pounds of product were
                 distributed.
  REASON       Products contain undeclared sulfites.
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-514-8.
  CODE         Unit numbers: 4799682, 4799560, 4760515,
                 4760139, 4760146, 4799585, 4799594, 4761333,
                 4761551.
  MANUFACTURER National Naval Medical Center, Bethesda,
                 Maryland.
  RECALLED BY  Manufacturer, by telephone on or about
                 February 5, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION New Jersey.
  QUANTITY     9 units were distributed.
  REASON       Blood products were not tested for HIV-1 Ag.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-618/619-8.
  CODE         Unit #1385890.
  MANUFACTURER Central Texas Regional Blood and Tissue
                 Center, Austin, Texas.
  RECALLED BY  Manufacturer, by letter dated September 23,
                 1997, and October 17, 1997.  Firm-initiated
                 recall complete.
  DISTRIBUTION Texas and Florida.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 with a history of exposure to hepatitis.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c)
                 Platelets, Pheresis.  Recall #B-645/647-8.
  CODE         Unit numbers:  a) 038E50493, 038K43442,
                 038K43521, 038K44340, 038FC18942, 038FC26037,
                 038FC26040, 038FC26399, 038FC26810
               b) 038E50493, 038K43442, 038K43521, 038FC26037
               c) 038P43542, 038P47121, 038P47190.
  MANUFACTURER American Red Cross Great Lakes Region, Fort
                 Wayne, Indiana.
  RECALLED BY  Manufacturer, by telephone on August 19, 1997,
                 and by letter dated September 24, 1997. 
               Firm-initiated recall complete.
  DISTRIBUTION Indiana and Ohio.
  QUANTITY     9 units; b) 4 units; c) 3 units were
                 distributed.
  REASON       Blood products were mislabeled as CMV antibody
                 negative, were distributed.  
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Cryoprecipitated AHF;
                 c) Plasma; d) Platelets, for Further
                 Manufacture; e) Recovered Plasma.
               Recall #B-657/661-8.
  CODE         Unit numbers:  a) 53FQ28625 and 53FQ10004
               b) 53FQ28625; c) 53FQ10004; d) 53FQ10004
               e) 53FQ28625.
  MANUFACTURER American Red Cross Blood Services, Baltimore,
                 Maryland.
  RECALLED BY  Manufacturer, by letters dated October 21 and
                 24, 1997, and December 4, 1997.  Firm-initiated recall complete.
  DISTRIBUTION Maryland, District of Columbia, Pennsylvania,
                 Switzerland.
  QUANTITY     a) 2 units; b) 1 unit; c) 1 unit; d) 1 unit;
                 e) 1 unit was distributed.
  REASON       Blood product were collected from a donor who
                 reported previously testing positive for
                 hepatitis.
  
  _______________
  PRODUCT      Source Leukocytes, For Manufacturing Use Only. 
                 Recall #B-693-8.
  CODE         Unit #01KK33558.
  MANUFACTURER American Red Cross Blood Services, Syracuse,
                 New York.
  RECALLED BY  Manufacturer, by fax on May 30, 1997.  Firm-initiated recall complete.
  DISTRIBUTION New Jersey.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor
                 taking the drug Methotrexate.
  
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Fresh Frozen Plasma; c)
                 Source Leukocytes; d) Recovered Plasma; e)
                 Platelets for further manufacture.
               Recall #B-830/834-8.
  CODE         Unit numbers:  a) 2194948, 1609413
               b) 1609413; c) 2194948; d) 2194948 
               e) 2194948.
  MANUFACTURER Central Indiana Regional Blood Center, Inc.,
                 Indianapolis, Indiana.
  RECALLED BY  Manufacturer, by letter on September 8 and 17,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Indiana, New Jersey, Switzerland.
  QUANTITY     a) 2 units; b-e) 1 unit of each component was
                 distributed.
  REASON       Blood products were collected from donors
                 taking the drug Proscar.
  
  _______________
  PRODUCT      Recovered Plasma.  Recall #B-
  -8.
  CODE         Unit #13FJ26037.
  MANUFACTURER American Red Cross Southeastern Regional,
                 Detroit, Michigan.
  RECALLED BY  Manufacturer, by fax dated January 14, 1998. 
                 Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor with
                 a history of cancer.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-854-8.
  CODE         Unit #13FM21536.
  MANUFACTURER American Red Cross Southeastern Region,
                 Detroit, Michigan.
  RECALLED BY  Manufacturer, by letter dated October 10,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Michigan.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 traveled to an area designated as endemic for
                 malaria.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets.
               Recall #B-856/857-8.
  CODE         Unit #38-34277.
  MANUFACTURER Michigan Community Blood Centers, Midland,
                 Michigan.
  RECALLED BY  Manufacturer, by letter on March 7, 1997. 
                 Firm-initiated recall complete.
  DISTRIBUTION Michigan.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 with a history of Hashimoto's Disease.
  
  _______________
  PRODUCT      a) Recovered Plasma; b) Platelets for Further
                 Manufacture.  Recall #B-861/862-8.
  CODE         Unit #18R86354.
  MANUFACTURER American Red Cross, Lansing, Michigan.
  RECALLED BY  Manufacturer, by letter or fax on September 8,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood products were collected from a donor
                 whose spouse tested positive for hepatitis C.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-865-8.
  CODE         Unit #0103837.
  MANUFACTURER Ashford Blood Bank, Inc., Ashford Medical
                 Center, Santruce, Puerto Rico.
  RECALLED BY  Manufacturer, by telephone on September 2,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Puerto Rico.
  QUANTITY     1 unit was distributed.
  REASON       Blood product tested initially reactive for
                 the antibody to the Hepatitis B core antigen
                 (HBc), with neither duplicate nor confirmatory
                 test performed.
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
  _______________
  PRODUCT      Fresh Frozen Plasma.  Recall #B-515-8.
  CODE         Unit #4764980.
  MANUFACTURER National Naval Medical Center, Bethesda,
                 Maryland.
  RECALLED BY  Manufacturer, by telephone on or about
                 February 5, 1997.  Firm-initiated recall
                 complete.
  DISTRIBUTION New Jersey.
  QUANTITY     1 unit was distributed.
  REASON       Blood products were not tested for HIV-1 Ag.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-560-8.
  CODE         Unit #17182-2574.
  MANUFACTURER United Blood Services, McAllen, Texas.
  RECALLED BY  Manufacturer, by telephone on September 17,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Texas.
  QUANTITY     1 unit was distributed.
  REASON       Blood product may have remained at room
                 temperature for an unacceptable length of
                 time.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-855-8.
  CODE         Unit #38FC31432.
  MANUFACTURER American Red Cross - Fort Wayne Region, Fort
                 Wayne, Indiana.
  RECALLED BY  Manufacturer, by telephone on December 19,
                 1997.  Firm-initiated recall complete.
  DISTRIBUTION Ohio.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was exposed to unacceptable
                 temperatures.
  
  
  RECALLS AND FIELD CORRECTIONS:  DEVICES --
  CLASS II =========
  _______________
  PRODUCT      a) VAPR Side Effect Electrode, Sterile,
                 Disposable, Catalog No. 225301
               b) VAPR Angled Side Effect Electrode, Sterile,
                 Disposable, Catalog No. 225302, used in
                 arthrosopic surgical procedures.
               Recall #Z-474/475-8.
  CODE         Model #25301, Lot numbers:  970432, 970722,
                 9705801, 9706801, 9707802 9708803, 9709801,
                 9709802, 9709804, 9709805,9709806,9709807 
               Model #225302, Lot numbers:  970433, 9707806,
                 970724, 9709803.
  MANUFACTURER Gyrus Medical Ltd., Cardiff, Wales, United
                 Kingdom.
  RECALLED BY  Mitek Products, Westwood, Massachusetts, by
                 letter on March 23, 1998.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION Nationwide, Austria, France, Italy, Spain
                 Japan, United Kingdom, Scotland.
  QUANTITY     a) 5677 units; b) 729 units were distributed.
  REASON       Electrode may crack and fragment resulting in
                 medical intervention.
  
  _______________
  PRODUCT      Suture Retrievers, packaged individually in
                 sterile Tyvek pouches, a component of the BTS
                 Mini-Laparotomy MMK Set, Model UM-250, and the
                 BTS Urethropexy Suture Placement Set, Model
                 UM-300, a sterile, disposable device (for
                 single patient use only) used to capture and
                 retrieve the suture tails during a
                 urethropexy, or an MMK surgical procedure,
                 included as a component of urethropexy sets,
                 and of mini-laparotomy MMK kits, and is not
                 sold separately. Recall #Z-480/481-8.
  CODE         Lot #9ML80206-04 (Model UM-250,
                 Mini-Laparoscopy MMK Set) and Lot #9ML71201-03
                 (Model UM-300, Urethropexy Kits).
  MANUFACTURER Louisville Laboratories, Inc., Louisville,
                 Kentucky.
  RECALLED BY  MedWorks Corporation, Louisville, Kentucky, by 
                 letter on March 6, 1998.  Firm-initiated
                 recall complete.
  DISTRIBUTION Pennsylvania, New Jersey, Virginia, Illinois,
                 Nevada, Oklahoma, Washington state.
  QUANTITY     56 surgical sets were distributed.
  REASON       Lack of assurance of sterility of the devices
                 due to possible puncture of the packaging by
                 sharp tip of Suture Retriever.
  
  
  RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
  _______________
  PRODUCT      Stone Extractor Baskets, used for endoscopic
                 removal of stones and foreign bodies.
               Recall #Z-423/473-8.
  CODE         1.  Product No. BMB5-2X4-8;
               2.  Product No. MB5-2X4-8;
               3.  Product No. MB5-2X4-8-MP3205;
               4.  Product No. MB5-3X6-8;
               5.  Product No. MBG-220-1.5X2.5;
               6.  Product No. MBG-220-2X4;
               7.  Product No. MPB-2X4-8;
               8.  Product No. MPB-3X6-8;
               9.  Product No. MSB-1.5X3-5;
               10. Product No. MSB-1.5X3.5-0169;
               11. Product No. MSB-1.5X3.5-0346;
               12. Product No. MSB-1.5X3.5-0463;
               13. Product No. MSB-1.5X3.5-F;
               14. Product No. MSB-1.5X3.5-F-0169;
               15. Product No. MSB-1.5X3.5-F-0463;
               16. Product No. MSB-1.5X3.5-SF;
               17. Product No. MSB-2.5X5; 
               18. Product No. MSB-2.5X5-0169;
               19. Product No. MSB-2.5X5-0346;
               20. Product No. MSB-2.5X5-0463;
               21. Product No. MSB-2.5X5-F;
               22. Product No. MSB-2.5X5-F-0169;
               23. Product No. MSB-2.5X5-F-0463;
               24. Product No. MSB-2.5X5-SF;
               25. Product No. MSB-2X4;
               26. Product No. MSB-2X4-0169;
               27. Product No. MSB-2X4-0346;
               28. Product No. MSB-2X4-0463;
               29. Product No. MSB-2X4-6;
               30. Product No. MSB-2X4-F;
               31. Product No. MSB-2X4-F-0169;
               32. Product No. MSB-2X4-F-0463;
               33. Product No. MSB-2X4-SF;
               34. Product No. MSB-3X6;
               35. Product No. MSB-3X6-0169;
               36. Product No. MSB-3X6-0346;
               37. Product No. MSB-3X6-0463;
               38. Product No. MSB-3X6-6;
               39. Product No. MSB-3X6-F;
               40. Product No. MSB-3X6-F-0169;
               41. Product No. MSB-3X6-F-0463;
               42. Product No. MSB-3X6-SF;
               43. Product No. MSB-60-2X4;
               44. Product No. MSB5-1.5X3.5;
               45. Product No. MSB5-2X4;
               46. Product No. MWB-1.5X3.5;
               47. Product No. MWB-2.5X5;
               48. Product No. MWB-2X4;
               49. Product No. MWB-3X6;
               50. Product No. MWB5-1.5x3.5;
               51. Product No. WCMB-200-4.
  MANUFACTURER Wilson-Cook Medical, Inc., Winston-Salem,
                 North Carolina.
  RECALLED BY  Manufacturer, by telephone and by mail on or
                 about November 5, 1997.  Firm-initiated recall
                 ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     1,363 units were distributed.
  REASON       The sheath may separate from the handle
                 rendering the device inoperable.  
  
  _______________
  PRODUCT      Roche Calibrator Serum a Class II device, is
                 an in vitro diagnostic calibrator intended for
                 use on COBAS chemistry instruments with Roche
                 clinical chemistry reagents.  Recall #Z-482-8.
  CODE         Product Code: 44157, Lot #s 0005 and 0005A EXP
                 12/1/98.
  MANUFACTURER Chiron Diagnostics, Irvine, California.
  RECALLED BY  Roche Diagnostic Systems, Somerville, New
                 Jersey, by letter on October 19, 1996.  
               Firm-initiated recall complete.
  DISTRIBUTION Nationwide, Israel, Australia, Switzerland.
  QUANTITY     1,987 kits (lot 0005) and 493 kits (lot 005A)
                 were distributed; firm estimates none remains
                 on the market.
  REASON       The device does not meet its labeling claim
                 for six days of stability after
                 reconstitution.
  
  _______________
  PRODUCT      Roche Control Serum and Calibrator:
               a) Control Serum N, a lyophilized preparation
                 of human serum used to monitor the accuracy
                 and precision of clinical chemistry assay
                 procedures; b) Control Serum P, a lyophilized
                 preparation of human serum used to monitor the
                 accuracy and precision of clinical chemistry
                 assay procedures; c) Calibrator (Human), a
                 human serum-based calibrator with lot specific
                 established values for the major constituents
                 of diagnostic interest commonly measured in
                 clinical laboratories.  Recall #Z-483/485-8.
  CODE         Lot numbers:  a) N3041, N3041A, N3041B;
               b) N3141, N3141A; c) T1237, T0941.
  MANUFACTURER Boehringer Mannheim, Mannheim, Germany.
  RECALLED BY  Roche Diagnostic Systems, Somerville, New
                 Jersey, by letter on March 31, 1997.  
                 Firm-initiated recall complete.
  DISTRIBUTION Nationwide and international.
  QUANTITY     Amount Shipped: 
               Control Serum N
               lot N3041   -  36 kits
               lot N3041A2 -  25 kits
               lot N3041B  -  62 kits
               Control Serum P
               lot N3141  -  25 kits
               lot N3141A -  67 kits
               Roche Calibrator
               lot T1237 - 100 kits
               lot T0941 -  24 kits.
  REASON       These control sera and calibrators were
                 calibrated against an incorrect standard and
                 yielded results that were 10-15 % higher than
                 expected.
  
  END OF ENFORCEMENT REPORT FOR APRIL 8, 1998.  BLANK PAGES MAY FOLLOW.

                                   ####
 
  