PRODUCT
The product is packaged in cases of 12/4 oz. plastic bottles; each
bottle in a paper board carton. Label information states: Gripe Water All Natural
Apple Flavor. An herbal supplement used to ease the gas and stomach discomfort
often associated with colic, hiccups and teething. Dietary Supplement. 4 fl.
oz. (120 ml), Recall # F-028-8
CODE
Unit: 26952V
RECALLING FIRM/MANUFACTURER
Recalling Firm: MOM Enterprises, Inc., San Rafael, CA, by telephone and
letter beginning on September 19, 2007.
Manufacturer: Botanical Laboratories Inc., Ferndale, WA. Firm initiated
recall is ongoing.
REASON
FDA confirmed through laboratory analysis of this lot of product the presence
of cryptosporidium. The product was sampled and tested after investigating
the illness of a 6-week-old infant in Minnesota who consumed the product. Cryptosporidium
is a parasite that can cause intestinal infections.
VOLUME OF PRODUCT IN COMMERCE
17,780 bottles
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
Jackson's Quality Plus Smoked Wild Salmon Spread Net Wt. 7 oz. Product
is in round plastic container, Recall # F-152-8
CODE
SELL BY dates up to and including 01/15/08 stamped on the side of the container.
SELL BY dates may have slashes between month/day/year; and may not. The firm
is not always consistent with their slashes.
RECALLING FIRM/MANUFACTURER
Quality Plus Products Inc., Bellingham, WA, by press release or telephone on
November 29, 2007 and by visit on December 5, 2007. Firm initiated recall is
complete.
REASON
Product labeling does not declare egg yolks as a sub-ingredient of horseradish.
VOLUME OF PRODUCT IN COMMERCE
120 cases
DISTRIBUTION
WA
___________________________________
PRODUCT
Del Rey White Flour Tortillas; all sizes; the tortillas are packaged 12 tortillas
per plastic bag, 40 packs per case for 6'' tortillas, 24 packs per case for
8'' tortillas, and 10 packs per case for 10'', 12'', 13'' and 14'' tortillas;
the tortillas are labeled as follows:
A) Del Rey Tortillas de Harina (6''), 12 Flour Tortillas, Net Wt. 12 Oz. (340g),
For Soft Shell Tacos, UPC 52771-20001;
B) Del Rey Tortillas de Harina (8''), 12 Flour Tortillas, Net Wt. 15 Oz. (425g),
Fajita Size, UPC 52771-20003;
C) Del Rey Burritos #2 (10''), 12 Tortillas, Net Wt. 27 oz. (765g), UPC 52771-40001;
D) Del Rey Burritos #3 (12''), 12 Tortillas, Net Wt. 34 oz. (964g), UPC 52771-00001;
E) Del Rey Burritos #3-1/2 (13''), 12 Tortillas, Net Wt. 42 oz. (1191 g), UPC
52771-50001 F) Del Rey Burritos #4 (14''), 12 Tortillas, Net Wt. 54 oz. (1.534
Kg), UPC 52771-50002, Recall # F-153-8
CODE
All sizes of flour tortillas with the Del Rey brand and use-by dates of NOV
11/2007 or earlier (SEP/17/2007, OCT/17/07, OCT/20/07, OCT/24/07, NOV/04/07,
NOV/10/07, NOV/11/07)
RECALLING FIRM/MANUFACTURER
Del Ret Tortilleria, Inc., Chicago, IL, by press release and letters dated
November 12, 2007 and by telephone or e-mail on November 13 and November 15,
2007. Firm initiated recall is complete.
REASON
Consumption of the flour tortillas has been associated with a series of food
borne illness outbreaks among individuals, mostly children, who complained
of stomach pains, vomiting, diarrhea, nausea, and headaches. These reported
symptoms typically occurred very soon after consuming the flour tortillas and
resolved within one day. There does not appear to be any long-term adverse
health effects.
VOLUME OF PRODUCT IN COMMERCE
8,784 cases
DISTRIBUTION
DE, FL, IL, IA, KS, ME, MN, NB, NH, NY, WI and Canada
___________________________________
PRODUCT
Hostess 100 Calorie Packs Chocolate Cake with Creamy Filling, 6-1.3 oz (37
g) packs Net Wt 7.75 oz. (220 g); 3 mini cakes per pack, 6 packs per box; UPC
0-45000-01114-4, Recall # F-151-8
CODE
Use By OCT 15 2007 00259A
RECALLING FIRM/MANUFACTURER
Recalling Firm: Interstate Brands Corp Hdqtrs, Kansas City, MO, by e-mail
on September 28, 2007.
Manufacturer: Interstate Brands Corp., Schiller Park, IL. Firm initiated
recall is ongoing.
REASON
The cakes were off-taste or tasted salty and were found to have levels of sodium
bicarbonate in excess of the firm's quality standards.
VOLUME OF PRODUCT IN COMMERCE
22,692 boxes
DISTRIBUTION
IL, IN, IA, KY, MI, MN, MO, OH, PA, WV and WI
___________________________________
PRODUCT
a) Buffy Coat, Recall # B-0371-08;
b) Red Blood Cells, Recall # B-0372-08;
c) Platelets, Recall # B-0373-08;
d) Fresh Frozen Plasma, Recall # B-0374-08;
e) Recovered Plasma, Recall # B-0375-08
CODE
a) FV16707;
b) FV17274, FV16707, FV16173, FV15720, FV15174, FV14583, KJ85579;
c) FV17274, FV15720, FV15174, FV14583, KJ85579;
d) FV17274, FV15174, KJ85579;
e) FV16173, FV15720, FV14583
RECALLING FIRM/MANUFACTURER
Central Blood Bank, Pittsburgh, PA, by facsimile on June 7, 2007. Firm initiated
recall is complete.
REASON
Blood products, which tested negative for Hepatitis C, but were collected from
a donor who previously tested positive for Hepatitis C, were distributed.
VOLUME OF PRODUCT IN COMMERCE
19 units
DISTRIBUTION
PA and FL
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0410-08
CODE
Unit: 49KQ13791
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by letter dated
July 20, 2005.
REASON
Blood product, collected from a donor who traveled to an area considered endemic
for malaria, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
TX
___________________________________
PRODUCT
Platelets Pheresis Leukocytes Reduced Irradiated, Recall # B-0419-08
CODE
Unit # 49KP05974
RECALLING FIRM/MANUFACTURER
Recalling Firm: American National Red Cross SW Region – HT, Tulsa, OK,
by telephone on December 8, 2005.
Manufacturer: American Red Cross Blood Services-SW Region, Farmers Branch,
TX. Firm initiated recall is complete.
REASON
Blood product, collected in a manner that may compromised the sterility of
the unit, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
OK
___________________________________
PRODUCT
a) Platelets, Recall # B-0499-08;
b) Recovered Plasma, Recall # B-0500-06
CODE
a) and b) Unit: 2607554
RECALLING FIRM/MANUFACTURER
Carter BloodCare, Bedford, TX, by telephone on June 24, 2005, and by letter
and electronic notification on July 22, 2005. Firm initiated recall is complete.
REASON
Blood products, collected from a donor for whom donor suitability was not adequately
determined, were distributed.
VOLUME OF PRODUCT IN COMMERCE
2 units
DISTRIBUTION
TX and Switzerland
___________________________________
PRODUCT
Red Blood Cells (Apheresis) Leukocytes Reduced, Recall # B-0511-08
CODE
Unit: 6278322 (2 units)
RECALLING FIRM/MANUFACTURER
Carter BloodCare, Bedford, TX, by facsimile on January 30, 2006. Firm initiated
recall is complete.
REASON
Blood products, collected from a donor who was at risk for variant Creutzfeldt-Jakob
disease (vCJD), were distributed.
VOLUME OF PRODUCT IN COMMERCE
2 units
DISTRIBUTION
TX
___________________________________
PRODUCT
a) Cryoprecipitated AHF, Recall # B-0512-08;
b) Red Blood Cells, Recall # B-0513-08;
c) Recovered Plasma, Recall # B-0514-08
CODE
a), b), and c) Unit: 71N65637X
RECALLING FIRM/MANUFACTURER
LifeSouth Community Blood Centers, Inc., Dunwoody, GA, by telephone and
letter dated December 22, 2005. Firm initiated recall is complete.
REASON
Blood products, collected from a donor who was at risk for HIV Group O, were
distributed.
VOLUME OF PRODUCT IN COMMERCE
3 units
DISTRIBUTION
GA and Switzerland
___________________________________
PRODUCT
Red Blood Cells, Leukocytes Reduced, Recall # B-0515-08
CODE
Unit: C77707
RECALLING FIRM/MANUFACTURER
South Bend Medical Foundation, South Bend, IN, by letter dated June 7, 2007.
Firm initiated recall is complete.
REASON
Blood product, which may have been contaminated with coagulase negative
Staphylococcus, were distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
IN
___________________________________
PRODUCT
a) Red Blood Cells Leukocytes Reduced, Recall # B-0516-08;
b) Fresh Frozen Plasma, Recall # B-0517-08;
c) Platelets Leukocytes Reduced, Recall # B-0518-08;
d) Red Blood Cells Leukocytes Reduced Irradiated, Recall # B-0519-08
CODE
a) Units: 49GW17421, 49GW17221, 49GW17229, 49GW17397, 49GW16791,
49GW16795, 49GW16807, 49GW17380, 49GW17395, 49GW17412, 49GW17203, 49GW17206,
49GW17207, 49GW17209, 49GW17211, 49GW17215, 49GW17217, 49GW17219, 49GW17225,
49GW17226, 49GW17231, 49GW17235, 49GV21062, 49GL15483, 49GW16794, 49GW16796,
49GW16800, 49GW16801, 49GW16803, 49GW16805, 49GW16809, 49GW16811, 49GW16684,
49GW16686, 49GW16689, 49GW16693, 49GW16695, 49GW16697, 49GW16698, 49GW16661,
49GW16664, 49GW16665, 49GW16669, 49GW16671, 49GW16673, 49GW16676, 49GW16639,
49GW16646, 49GW16649, 49GW16651, 49GL15420, 49GL15421, 49GL15423, 49GL15426,
49GL15427, 49GL15432, 49GL15434, 49GL15439, 49GL15440, 49GL15441, 49GN31362,
49GN31373, 49GN31376, 49GN31379, 49GN31381, 49GN31383, 49GN31385, 49GN31388,
49GN31390, 49GW16520, 49GW16522, 49GW16523, 49GW16524, 49GW16530, 49GW16539,
49GC10805, 49GC10807, 49GC10811, 49GC10817, 49GC10823, 49GM11759, 49GM11761,
49GM11763, 49GM11768, 49GV20575, 49GV20583, 49GV20604, 49GV20622, 49GV20628,
49GV20651, 49GV20490, 49GN30699, 49GN30678;
b) Units: 49GW17448, 49GW17405, 49GW17409, 49GW17421, 49GW17221, 49GW16791,
49GW17382, 49GW17412, 49GW17428, 49GW17430, 49GW17203, 49GW17206, 49GW17207,
49GW17215, 49GW17217, 49GW17219, 49GW17225, 49GW17226, 49GW17231, 49GW17235,
49GW16794, 49GW16684, 49GW16697, 49GW16698, 49GW16661, 49GW16669, 49GW16673,
49GW16676, 49GW16639, 49GW16646, 49GW16651, 49GL15420, 49GL15421, 49GL15423,
49GL15426, 49GL15427, 49GL15432, 49GL15434, 49GL15439, 49GL15440, 49GL15441,
49GN31362, 49GN31376, 49GN31379, 49GN31381, 49GN31385, 49GN31388, 49GW16520,
49GW16523, 49GW16524, 49GW16527, 49GW16530, 49GW16535, 49GW16537, 49GW16539,
49GC10805, 49GC10807, 49GC10811, 49GC10817, 49GC10823, 49GM11759, 49GM11761,
49GM11763, 49GM11768, 49GV20575, 49GV20583, 49GV20622, 49GV20628, 49GV20490,
49GN30698, 49GN30699, 49GN30678;
c) Units: 49GN32338, 49GN32339, 49GW17405, 49GW17409, 49GW17421, 49GW17221,
49GW17229, 49GM12123, 49GW17386, 49GW17391, 49GW17392, 49GW17403, 49GW17404,
49GW17408, 49GW17412, 49GW17413, 49GW17420, 49GW17424, 49GW17428, 49GW17203,
49GW17206, 49GW17207, 49GW17215, 49GW17217, 49GW17226, 49GW16693, 49GW16695,
49GW16697, 49GW16673, 49GW16639, 49GW16646, 49GL15423, 49GL15426, 49GL15434,
49GL15439, 49GL15440, 49GL15441, 49GN31388, 49GW16520, 49GW16530, 49GW16535,
49GW16537, 49GW16539, 49GC10823, 49GM11759, 49GM11768, 49GV20575, 49GV20583,
49GN30678, 49GW17380, 49GW17219, 49GW16684, 49GW16651;
d) Unit: 49GW16679
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by telephone
beginning June 25, 2004. Firm initiated recall is complete.
REASON
Blood products, collected in a manner that may have compromised the sterility
of the units, were distributed.
VOLUME OF PRODUCT IN COMMERCE
219 units
DISTRIBUTION
OK and TX
___________________________________
PRODUCT
a) Red Blood Cells Leukocytes Reduced, Recall # B-0520-08;
b) Fresh Frozen Plasma, Recall # B-0521-08;
c) Platelets Leukocytes Reduced, Recall # B-0522-08;
d) Red Blood Cells Leukocytes Reduced Irradiated, Recall # B-0523-08;
CODE
a) Units: 49GN36384, 49GN36372, 49GN36369, 49GN36362, 49GN36359, 49GW20980,
49GN36702, 49GN36713, 49GN36715, 49GN36718, 49GN36732, 49GN36737, 49GN36739,
49GN36742, 49GV23212, 49GV23211, 49GV23203, 49GV23202, 49GV23200, 49GV23198,
49GV23197, 49GV23194, 49GV23188, 49GV23182, 49GV23179, 49GN36989, 49GV36980,
49GM13778, 49GM13774, 49GM13771, 49GM13768, 49GW21559, 49GW21558, 49GW21550,
49GW21547, 49GW21545, 49GQ53350, 49GQ53329, 49GQ53315, 49GQ53295, 49GL17490,
49GL17489, 49GL17469, 49GL17465, 49GL17462, 49GW21969, 49GW21967, 49GW21965,
49GW21961, 49GM14305, 49GM14313, 49GM14318, 49GM14320, 49GM14286, 49GM14291,
49GM14295, 49GM14298, 49GV23962, 49GV23975, 49GN38764, 49GN38766, 49GN38818;
b) Units: 49GN36370, 49GN36369, 49GN36362, 49GN36359, 49GW20980, 49GN36702,
49GN36735, 49GN36737, 49GN36742, 49GV23212, 49GV23211, 49GV23203, 49GV23202,
49GV23200, 49GV23198, 49GV23197, 49GV23182, 49GV23179, 49GV36980, 49GM13779,
49GM13778, 49GM13774, 49GM13771, 49GM13768, 49GW21559, 49GW21545, 49GQ53350,
49GL17489, 49GL17490, 49GL17462, 49GW21967, 49GW21963, 49GM14313, 49GM14318,
49GM14921, 49GV23938, 49GV23943, 49GV23966, 49GV23975, 49GN38764;
c) 49GN36370, 49GN36369, 49GN36362, 49GN36359, 49GN36702, 49GN36732, 49GN36735,
49GN36737, 49GN36742, 49GV23212, 49GV23203, 49GV23200, 49GV23182, 49GV23179,
49GW21559, 49GW21550, 49GW21547, 49GW21545, 49GQ53350, 49GQ53315, 49GL17490,
49GL17489, 49GW21969, 49GW21967, 49GW21963, 49GW21961, 49GM14291, 49GM14298,
49GV23975, 49GN38764, 49GN38818;
d) 49GN36370, 49GN36735, 49GV23206, 49GV23938, 49GV23966
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by telephone
beginning February 2, 2005. Firm initiated recall is complete.
REASON
Blood products, collected in a manner that may have compromised the sterility
of the units, were distributed.
VOLUME OF PRODUCT IN COMMERCE
138 units
DISTRIBUTION
OK, TX, NV, UT
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0553-08
CODE
Unit: 49KM17577
RECALLING FIRM/MANUFACTURER
Recalling Firm: American National Red Cross SW Region – HT, Tulsa, OK,
by facsimile or electronic notification on March 23, 2007.
Manufacturer: American Red Cross Blood Services-SW Region, Farmers Branch, TX.
Firm initiated recall is complete.
REASON
Blood product, collected from a donor for whom donor suitability was not adequately
determined, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
TX
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0555-08
CODE
Unit: 49N57219
RECALLING FIRM/MANUFACTURER
Recalling Firm: American National Red Cross SW Region – HT, Tulsa, OK,
by telephone, facsimile and electronic notification on March 23, 2007.
Manufacturer: American Red Cross Blood Services-Southwest Region, Harlingen, TX.
Firm initiated recall is complete.
REASON
Blood product, collected from a donor for whom donor suitability was not adequately
determined, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
TX
___________________________________
PRODUCT
Platelets Pheresis Leukocytes Reduced, Recall # B-0557-08
CODE
Unit: 49KP10021
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by telephone
on June 19, 2007 and by facsimile on June 21, 2007. Firm initiated recall
is complete.
REASON
Blood product, which was inappropriately distributed in a single storage bag,
was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
TX
___________________________________
PRODUCT
Platelets Pheresis Leukocytes Reduced, Recall # B-0558-08
CODE
Unit: 8607814 (part 1)
RECALLING FIRM/MANUFACTURER
Lifesource, Glenview, IL, by telephone on October 2, 2007. Firm initiated recall
is complete.
REASON
Blood product, which was labeled as leukoreduced, but was not tested to verify
the White Blood Cell count, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
IL
___________________________________
PRODUCT
Red Blood Cells, Recall # B-0559-08
CODE
Unit: 9422860
RECALLING FIRM/MANUFACTURER
Recalling Firm: Lifesource, Glenview, IL, by telephone on September 25, 2007.
Manufacturer: Lifesource Portage Park, Chicago, IL. Firm initiated recall is
complete.
REASON
Blood product, collected from a donor who lived in an area considered endemic
for malaria, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
IL
___________________________________
PRODUCT
Platelets, Recall # B-0560-08
CODE
Unit: 7567127
RECALLING FIRM/MANUFACTURER
Lifesource, Glenview, IL, by telephone on August 31, 2007. Firm initiated recall
is complete.
REASON
Blood product, collected from a donor who traveled in an area considered endemic
for malaria, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
IL
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0561-08
CODE
Unit: 49GV36784
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by facsimile
on December 12, 2006. Firm initiated recall is complete.
REASON
Blood product, collected from a donor for whom donor suitability was not adequately
determined, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
OK
___________________________________
PRODUCT
Crotalidae Polyvalent Immune Fab (Ovine), Recall # B-0318-08
CODE
Unit: BN200623
RECALLING FIRM/MANUFACTURER
Recalling Firm: Protherics Inc., Brentwood, TN, by letter on October 29,
2007.
Manufacturer: Protherics U.K. Limited, Llandysul, Ceredigion, UK. Firm initiated
recall is ongoing.
REASON
The product tested subpotent ate the 24-month time point.
VOLUME OF PRODUCT IN COMMERCE
1877 cartons
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0501-08
CODE
Unit: 2257864
RECALLING FIRM/MANUFACTURER
Central Kentucky Blood Center, Inc., Lexington, KY, by visit on January 6,
2006. Firm initiated recall is complete.
REASON
Blood product, which was not quarantined after receiving post donation information
concerning an illness at the time of donation, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
KY
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced Irradiated, Recall # B-0502-08
CODE
Units: 5911456, 5933035
RECALLING FIRM/MANUFACTURER
Carter BloodCare, Bedford, TX, by letter dated January 25, 2006. Firm initiated
recall is complete.
REASON
Blood products, which were not returned to controlled storage within eight
hours following collection, were distributed.
VOLUME OF PRODUCT IN COMMERCE
2 units
DISTRIBUTION
TX
___________________________________
PRODUCT
a) Platelets, Recall # B-0506-08;
b) Platelets Pooled Leukocytes Reduced, Recall # B-0507-08;
c) Platelets Pheresis Leukocytes Reduced, Recall # B-0508-08;
d) Red Blood Cells Leukocytes Reduced Irradiated, Recall # B-0509-08;
e) Red Blood Cells Leukocytes Reduced, Recall
# B-0510-08
CODE
a) Units: GR44481, GR44484, GR44490, GR44502, KC15269, and KC15275;
b) Units: GR44489, GR44509, GR44485, GR44508, GR44506, GR44518,
GR44495,
FQ55665, LE57544, GR44491, GR44507, LJ00379, GR44483, KC15050,
GR44364,
GR44379, LE57471, KC15129, LE57455, LE57445, LE57450, and
LE57422;
c) Units: KZ86759, LC35416, FQ55483, and LC35408;
d) Units: FW60443, FW60443, FW60443, FW60443, FW60443, FW60443, FW60443,
FW60443, FW60443, FR58564, FR58564, FR58564, FR58564, FR58564,
FR58564,
FR58564, FR58564, and FR58564;
e) Units: FJ95170, J95170, FJ95170, FQ55474, FQ55474, LC35541, LC35541, and
LC35541
RECALLING FIRM/MANUFACTURER
Blood Center of Wisconsin, Inc., Milwaukee, WI, by facsimile on June 20, 2003.
Firm initiated recall is complete.
REASON
Blood products, which may have not met acceptable product specifications, were
distributed.
VOLUME OF PRODUCT IN COMMERCE
58 units
DISTRIBUTION
WI
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0527-08
CODE
Units: 49GG20868, 49KQ14520, 49KQ14522, 49KQ14529
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by telephone
and facsimile beginning July 22, 2005. Firm initiated recall is complete.
REASON
Blood products, collected from donors who did not have a complete health history
screening, were distributed.
VOLUME OF PRODUCT IN COMMERCE
4 units
DISTRIBUTION
OK and TX
___________________________________
PRODUCT
Recovered Plasma, Recall # B-0554-08
CODE
Unit: 49KM17577
RECALLING FIRM/MANUFACTURER
Recalling Firm: American National Red Cross SW Region – HT, Tulsa, OK,
by facsimile or electronic notification on March 23, 2007.
Manufacturer: American Red Cross Blood Services-SW Region, Farmers Branch, TX.
Firm initiated recall is complete.
REASON
Blood product, collected from a donor for whom donor suitability was not adequately
determined, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
CA
___________________________________
PRODUCT
Recovered Plasma, Recall # B-0556-08
CODE
Unit: 49N57219
RECALLING FIRM/MANUFACTURER
Recalling Firm: American National Red Cross SW Region – HT, Tulsa, OK,
by telephone, facsimile and electronic notification on March 23, 2007.
Manufacturer: American Red Cross Blood Services-Southwest Region, Harlingen, TX.
Firm initiated recall is complete.
REASON
Blood product, collected from a donor for whom donor suitability was not adequately
determined, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
CA
___________________________________
PRODUCT
Red Blood Cells Leukocytes Reduced, Recall # B-0562-08
CODE
Unit: 49KQ14131
RECALLING FIRM/MANUFACTURER
American National Red Cross SW Region – HT, Tulsa, OK, by telephone
on October 19, 2005 and by facsimile on October 20, 2005. Firm initiated recall
is complete.
REASON
Blood product, collected from a donor for whom donor suitability was not adequately
determined, was distributed.
VOLUME OF PRODUCT IN COMMERCE
1 unit
DISTRIBUTION
TX
___________________________________
PRODUCT
Red Blood Cells (Apheresis) Leukocytes Reduced, Recall # B-0623-08
CODE
Units: 1642581( Parts 1 and 2), 1669201 (Parts 1 and 2)
RECALLING FIRM/MANUFACTURER
Hoxworth Blood Center UC Medical Center, Cincinnati, OH, by letter dated
June 20, 2006. Firm initiated recall is complete.
REASON
Blood products, collected from a donor who was taking the medication Enbrel,
were distributed.
VOLUME OF PRODUCT IN COMMERCE
4 units
DISTRIBUTION
OH
___________________________________
PRODUCT
Datascope, Panorama Patient Monitoring Network; Panorma Telepack 608; Software
Versions 8.1.X, 8.3, 8.3.1, 8.4.1, and 8.5.X, Part Number: 0998-00-0191-04,
Recall # Z-0150-2008
CODE
Part Number: 0998-00-0191-04
RECALLING FIRM/MANUFACTURER
Datascope Corporation, Fairfieldm, NJ,
by Urgent Product Field Correction notices on October 9, 2007.
REASON
Display problems: If an ECG cable, which has been damaged due to the ingress
of liquid or by mechanical trauma, is utilized with the Telepack, it may cause
the Telepack to switch between the 3-lead and 5 lead-input modes. During the
switching process, the digital heart rate displayed at the Panorama Central Station
will be frozen, and subsquent arrthymia alarms will not be announced.
VOLUME OF PRODUCT IN COMMERCE
338 sites
DISTRIBUTION
Nationwide
___________________________________
PRODUCT
a) Actiwatch 16 Part Number 198-0300-00, Actiwatch 16 AM 198-0300-AM;
Actiwatch 64 Part Number 198-0301-00; Actiwatch 64 AM Part
Number
198-0301-AM, Recall # Z-0324-2008;
b) Actical with Step, Part Number 198-0302-00, Recall # Z-0325-2008;
c) Actiwatch 64 AM, Bandless Part Number 198-0301-AM1, Recall # Z-0326-2008;
d) Actiwatch 16, Basic Startup pkg 198-1010-00; and Actiwatch 64, Basic Startup
pkg Part Number 198-1000-00, Recall # Z-0327-2008
CODE
a) Serial Numbers: V960641, V960642, V960655, V960656, V960657, V960658, V960660,
V960661, V960986, V961113, V961114, V961115, V961116, V961117, V961118, V961119,
V961120, V961177, V961178, V961179, V961180, V961181, V961182, V961183, V961184,
V961185, V961186, V961187, V961188, V961189, V961191, V961192, V961193, V961194,
V961195, V961196, V961197, V961198, V961199, V961200, V961513, V961514, V961515,
V961516, V961517, V961518, V961519, V961520, V961593, V961594, V961595, V961596,
V961597, V961598, V961599, V961600, V961601, V961602, V961603, V961604, V961605,
V961606, V961607, V961608, V961609, V961610, V961611, V961612, V961613, V961614,
V961615, V961616, V961617, V961121, V961122, V961123, V961124, V961125, V961126,
V961127, V961128, V961321, V961322, V961323, V961324, V961325, V961326, V961327,
V961328, V961329, V961330, V961331, V961332, V961333, V961334, V961335, V961336,
V961337, V961338, V961339, V961340,V960611, V960617, V960618, V960619, V960620,
V960621, V960622, V960629, V960630, V960631, V960632, V960635, V960638, V960639,
V960640, V960643, V960644, V960645, V960663, V960669, V960671, V960673, V960678,
V960681, V960682, V960684, V960685, V960686, V960689, V960690, V960692, V960693,
V960694, V960695, V960696, V960697, V960698, V960701, V960702, V960709, V960710,
V960720, V960721, V960722, V960723, V960724, V960725, V960726, V960728, V960729,
V960730, V960731, V960732, V960736, V960737, V960742, V960743, V960744, V960745,
V960746, V960747, V960748, V960749, V960751, V960752, V960753, V960754, V960755,
V960757, V960758, V960760, V960761, V960834, V960835, V960836, V960837, V960838,
V960839, V960840, V960841, V960842, V960843, V960844, V960845, V960846, V960847,
V960848, V960849, V960850, V960851, V960854, V960855, V960856, V960857, V960858,
V960859, V960860, V960861, V960862, V960863, V960871, V960872, V960875, V960876,
V960877, V960878, V960879, V960882, V960889, V960891, V960892, V960893, V960894,
V960895, V960896, V960897, V960898, V960900, V960902, V960903, V960904, V960913,
V960914, V960915, V960916, V960917, V960918, V960919, V960920, V960921, V960922,
V960923, V960924, V960925, V960928, V960929, V960930, V960931, V960932, V960933,
V960934, V960935, V960936, V960937, V960938, V960939, V960940, V960941, V960942,
V960943, V960944, V960945, V960946, V960947, V960948, V960949, V960950, V960951,
V960952, V960953, V960954, V960955, V960956, V960957, V960958, V960959, V960960,
V960961, V960962, V960963, V960964, V960965, V960966, V960967, V960968, V960969,
V960970, V960971, V960972, V960973, V960974, V960975, V960976, V960977, V960978,
V960979, V960980, V960981, V960982, V960983, V960984, V960993, V961009, V961016,
V961017, V961018, V961021, V961023, V961024, V961025, V961026, V961027, V961028,
V961029, V961030, V961032, V961034, V961035, V961036, V961038, V961039, V961040,
V961042, V961043, V961044, V961045, V961047, V961048, V961049, V961050, V961051,
V961052, V961053, V961054, V961056, V961057, V961058, V961059, V961060, V961064,
V961066, V961067, V961068, V961130, V961134, V961144, V961145, V961146, V961147,
V961148, V961149, V961150, V961151, V961152, V961201, V961202, V961203, V961204,
V961205, V961206, V961207, V961208, V961209, V961210, V961211, V961212, V961213,
V961214, V961215, V961216, V961260, V961278, V961283, V961320, V961348, V961351,
V961352, V961353, V961354, V961355, V961356, V961357, V961358, V961360, V961367,
V961368, V961369, V961370, V961371, V961372, V961373, V961374, V961375, V961376,
V961377, V961378, V961379, V961381, V961382, V961383, V961388, V961389, V961390,
V961391, V961392, V961393, V961394, V961396, V961397, V961398, V961399, V961400,
V961401, V961402, V961403, V961404, V961405, V961406, V961407, V961408, V961409,
V961410, V961411, V961412, V961413, V961414, V961415, V961416, V961417, V961418,
V961419, V961420, V961421, V961422, V961423, V961424, V961425, V961426, V961427,
V961428, V961429, V961430, V961431, V961432, V961433, V961434, V961435, V961436,
V961437, V961438, V961439, V961440, V961442, V961444, V961445, V961446, V961447,
V961448, V961449, V961450, V961451, V961452, V961453, V961454, V961455, V961456,
V961457, V961458, V961459, V961460, V961461, V961462, V961463, V961464, V961465,
V961466, V961467, V961468, V961469, V961470, V961471, V961472, V961473, V961474,
V961475, V961476, V961477, V961478, V961479, V961480, V961481, V961485, V961486,
V961487, V961488, V961510, V961511, V961512, V961521, V961522, V961523, V961524,
V961525, V961526, V961529, V961530, V961531, V961532, V961534, V961535, V961536,
V961538, V961539, V961540, V961542, V961543, V961544, V961545, V961546, V961547,
V961548, V961549, V961550, V961551, V961552, V961553, V961554, V961555, V961556,
V961561, V961591, V961592, V961625, V961626, V961627, V961628, V961629, V961630,
V961631, V961632, V961633, V961634, V961635, V961636, V961637, V961638, V961639,
V961640, V961641, V961642, V961643, V961644, V961645, V961647, V961648, V961649,
V961650, V961652, V961653, V961654, V961655, V961656, V961689, V961690, V961691,
V961692, V961693, V961694, V961695, V961696, V961697, V961698, V961699, V961700,
V961701, V961702, V961703, V961704, V961705, V961706, V961707, V961708, V961713,
V961714, V961715, V961716, V961717, V961718, V961727, V961728, V961729, V961730,
V961731, V961732, V961733, V961735, V961736, V961737, V961738, V961739, V961740,
V961741, V961742, V961896, V961897, V961900, V961901, V961902, V961903, V961904,
V961905, V961906, V961907, V961908, V961909, V961910, V961569, V961570, V961571,
V961572, V961573, V961574, V961575, V961576, V961752, V961753, V961754, V961755,
V961756, V961757, V961758, V961759, V961760, V961761, V961762, V961763, V961764,
V961765, V961766, V961768, V961769, V961770, V961771, and V961772;
b) Serial Numbers: B100001, B100002, B100003, B100005, B100006, B100007, B100008,
B100009, B100023, B100024, B100025, B100026, B100027, B100028, B100029, B100030,
B100031, B100032, B100033, B100034, B100035, B100036, B100037, B100038, B100039,
B100041, B100042, B100043, B100044, B100045, B100046, B100047, B100048, B100049,
B100050, B100051, B100052, B100053, B100054, B100055, B100056, B100057, B100058,
B100059, B100060, B100061, B100062, B100063, B100064, B100065, B100066, B100067,
B100068, B100069, B100070, B100071, B100072, B100073, B100074, B100075, B100076,
B100077, B100078, B100079, B100080, B100081, B100082, B100083, B100084, B100085,
B100086, B100087, B100088, B100089, B100090, B100091, B100092, B100093, B100094,
B100095, B100096, B100097, B100098, B100099, B100100, B100101, B100102, B100103,
B100104, B100105, B100106, B100107, B100108, B100109, B100110, B100111, B100112,
B100113, B100114, B100115, B100116, B100117, B100118, B100119, B100120, B100121,
B100122, B100123, B100124, B100125, B100126, B100127, B100128, B100129, B100130,
B100131, B100132, B100133, B100134, B100135, B100136, B100137, B100138, B100139,
B100140, B100141, B100142, B100143, B100144, B100145, B100146, B100147, B100148,
B100149, B100150, B100151, B100152, B100153, B100154, B100155, B100156, B100157,
B100158, B100159, B100160, B100161, B100162, B100163, B100164, B100165, B100166,
B100167, B100168, B100169, B100170, B100171, B100172, B100173, B100174, B100175,
B100176, B100177, B100178, B100179, B100180, B100181, B100182, B100183, B100184,
B100185, B100186, B100187, B100188, B100189, B100190, B100191, B100192, B100193,
B100194, B100195, B100196, B100197, B100198, B100199, B100200, B100201, B100202,
B100203, B100204, B100205, B100206, B100207, B100208, B100209, B100210, B100211,
B100212, B100213, B100214, B100215, B100216, B100217, B100218, B100219, B100220,
B100221, B100222, B100223, B100224, B100225, B100226, B100227, B100228, B100229,
B100230, B100231, B100232, B100233, B100234, B100235, B100236, B100237, B100238,
B100239, B100240, B100241, B100242, B100243, B100244, B100245, B100246, B100247,
B100248, B100249, B100250, B100251, B100252, B100253, B100254, B100255, B100256,
B100257, B100258, B100259, B100260, B100261, B100262, B100263, B100264, B100265,
B100266, B100267, B100268, B100269, B100270, B100271, B100272, B100273, B100274,
B100275, B100276, B100277, B100278, B100279, B100280, B100281, B100282, B100283,
B100284, B100285, B100286, B100287, B100288, B100289, B100290, B100291, B100292,
B100293, B100294, B100295, B100296, B100297, B100298, B100299, B100300, B100301,
B100302, B100303, B100304, B100305, B100306, B100307, B100308, B100309, B100310,
B100311, B100312, B100313, B100314, B100315, B100316, B100317, B100318, B100319,
B100320, B100321, B100322, B100323, B100324, B100325, B100326, B100327, B100328,
B100329, B100330, B100331, B100332, B100333, B100334, B100335, B100336, B100337,
B100338, B100339, B100340, B100341, B100342, B100343, B100344, B100345, B100346,
B100347, B100348, B100349, B100350, B100351, B100352, B100353, B100354, B100355,
B100356, B100357, B100358, B100359, B100360, B100361, B100362, B100363, B100364,
B100365, B100366, B100367, B100368, B100369, B100370, B100371, B100372, B100373,
B100374, B100375, B100376, B100377, B100378, B100379, B100380, B100381, B100382,
B100383, B100384, B100385, B100386, B100387, B100388, B100389, B100390, B100391,
B100392, B100393, B100394, B100395, B100396, B100397, B100398, B100399, B100400,
B100401, B100402, B100403, B100406, B100407, B100408, B100409, B100410, B100411,
B100412, B100413, B100414, B100415, B100416, B100417, B100418, B100419, and B100420;
c) Serial Numbers: V960132, V960133, V960135, V960191, V960253, V960254, V960255,
V960256, V960257, V960314, V960460, V960461, V960469, V960481, V960483, V960487,
V960488, V960504, V960505, and V960550;
d) Serial Numbers: V960700, V960703, V960735, V960739, V960740, V960864, V960865,
V960866, V960868, V960869, V960870, V961013, V961014, V961015, and
V960659
RECALLING FIRM/MANUFACTURER
Mini-Mitter Company, Inc., Bend, OR, by letter dated June 26, 2007. Firm initiated
recall is ongoing.
REASON
Communication Failure: the devices have a memory chip that has a limited number
of read/write cycles which can trigger set up information and/or subject information
anomalies or intermittent communication annomalies. If left unaddressed, the
issues can lead to permanent communication failure.
VOLUME OF PRODUCT IN COMMERCE
1,073 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
SERFAS Energy Suction Probe, Model 279-351-100, 3.5mm 90-S, Recall # Z-0484-2008
CODE
Lot Numbers: 06361AE2 06362AE2 07004AE2 07008AE2 07010AE2 07012AE2 07016AE2 07017AE2
07018AE2 07022AE2 07025AE2 07026AE2 07029AE2 07031AE2 07036AE2 07038AE2 07043AE2,
and 07044AE2
RECALLING FIRM/MANUFACTURER
Recalling Firm: Stryker Endoscopy, San Jose, CA, by recall notice dated
April 27, 2007.
Manufacturer: Stryker Puerto Rico, Inc., Arroyo, PR. Firm initiated recall
is ongoing.
REASON
Device Shut off/on: Device may not turn off after taking finger/foot off of activating
switch and may activate without pushing a button. (Patient burns may result from
these device malfunctions)
VOLUME OF PRODUCT IN COMMERCE
6,037 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
GE Medical Systems Information Technologies Clinical Information Center (CIC)
Pro v5.0.3 & v5.0.6 systems, GE Healthcare, Clinical Systems, Recall # Z-0519-2008
CODE
JA104480715GA JA104521071GA JA106405918GA JA106405954GA JA106426089GA JA106426106GA
JA104290048GA JA104389244GA JA104480664GA JA104480715GA JA105111564GA JA105121632GA
JA105151746GA JA105151749GA JA105171811GA JA105342546GA JA1054833330GA JA105493389GA
JA105493412GA JA1060043963GA JA10606491GA JA106074297GA JA106134601GA JA1061446747GA
JA106164772GA JA106245153GA JA106295351GA JA106355517GA JA106365600GA JA106426063GA
JA106426100GA JA106486211GA JA106486215GA JA107096739GA JA107096742GA JAI06345514GA
SCH06430010GA SCH06440050GA SCH06440062GA SCH06460310GA SCH06470347GA SCH06470350GA
SCH06510406GA SCH07040470GA SCH07050547GA SCH0706057GA SCH07070604GA SCH07070653GA
SCH07080669GA. SCH07080671GA SCH07080699GA SCH07080709GA SCH07080731GA SCH07120796GA
SCH07120809GA SCH07120825GA SCH07130839GA SCH07130865GA SCH07130882GA SCH07140933GA
SCH07150953GA SCH07150963GA SCH07161062GA SCH07181099GA SCH07191148GA SCH07191156GA.
SCH07201187GA SCH07201190GA SCH07211226GA SCH07231266GA SCH07231276GA SCH07231291GA
SCH07231303GA SCH07261336GA SCH07261339GA SCH07261351GA SCH07261354GA SCH07301566GA
SCH073115892GA SCH07321745GA SCH07331762GA SCH07331774GA SCH07331792GA SCH07331810GA
SCH07331811GA SCH07361897GA SCH07361902GA SCH07361911GA SCH07361931GA SCH07361956GA
SCH07362006GA SCH07382330GA SCH07080714GA SCH07392353GA SCH07161009GA JA104380193GA
SCH06240003GA SCH06430007GA SCH06430009GA SCH06430030GA SCH06430035GA SCH06430036GA
SCH06430039GA SCH06430041GA SCH06430045GA SCH06430048GA SCH06440049GA SCH06440051GA
SCH06440052GA SCH06440053GA SCH06440056GA SCH06440058GA SCH06440060GA SCH06440063GA
SCH06440064GA SCH06440065GA SCH06440067GA SCH06440069GA SCH06440072GA SCH06440076GA
SCH06440077GA SCH06440080GA SCH06440082GA SCH06440083GA SCH06440084GA SCH06440094GA
SCH06440117GA SCH06440119GA SCH06440121GA SCH06440122GA SCH06440123GA SCH06440126GA
SCH06440128GA SCH06440136GA SCH06440144GA SCH06440151GA SCH06440155GA SCH06440161GA
SCH06440175GA SCH06440184GA SCH06440185GA SCH06440187GA SCH06440190GA SCH06440202GA
SCH06450208GA SCH06450209GA SCH06450210GA SCH06450214GA SCH06450215GA SCH06450217GA
SCH06450219GA SCH06450223GA SCH06450227GA SCH06450232GA SCH06470352GA SCH06510388GA
SCH06510389GA SCH06510390GA SCH06510391GA SCH06510398GA SCH06510409GA SCH06510410GA
SCH06510412GA SCH06510414GA SCH06510417GA SCH06510418GA SCH06510419GA SCH06510421GA
SCH06510426GA SCH06510427GA SCH06510428GA SCH06510429GA SCH06510430GA SCH06510431GA
SCH07040496GA SCH07050506GA SCH07050520GA SCH07050556GA SCH07060590GA SCH07070610GA
SCH07070626GA SCH07110764GA SCH07110777GA SCH07120806GA SCH07120830GA SCH07140890GA
SCH07140894GA SCH07140904GA SCH07140906GA SCH07140912GA SCH07140922GA SCH07140936GA
SCH07140938GA SCH07140942GA SCH07150948GA SCH07150960GA SCH07150994GA SCH07151000GA
SCH07151003GA SCH07151006GA SCH07161019GA SCH07161025GA SCH07161054GA SCH07161062GA
SCH07161063GA SCH07221233GA SCH07221244GA SCH07261343GA SCH07291492GA SCH07291511GA
SCH07291513GA SCH07301516GA SCH07301519GA SCH07301522GA SCH07301523GA SCH07301524GA
SCH07301525GA SCH07301529GA SCH07301531GA SCH07301532GA SCH07301534GA SCH07301535GA
SCH07301540GA SCH07301541GA SCH07301542GA SCH07301543GA SCH07301544GA SCH07301545GA
SCH07301548GA SCH07301550GA SCH07301551GA SCH07301553GA SCH07301554GA SCH07301555GA
SCH07301556GA SCH07301558GA SCH07301560GA SCH07301562GA SCH07301564GA SCH07301565GA
SCH07301566GA SCH07301567GA SCH07301568GA SCH07301569GA SCH07301571GA SCH07301572GA
SCH07301573GA SCH07301574GA SCH07301575GA SCH07301577GA SCH07301580GA SCH07301581GA
SCH07301582GA SCH07311583GA SCH07311586GA SCH07311587GA SCH07311591GA SCH07311592GA
SCH07311593GA SCH07311597GA SCH07311608GA SCH07311611GA SCH07311613GA SCH07321667GA
SCH07321686GA SCH07321694GA SCH07321698GA SCH07321699GA SCH07321721GA SCH07331766GA
SCH07341852GA SCH07361897GA SCH07361927GA SCH07362014GA SCH07362027GA SCH07362046GA
SCH07362047GA SCH07362048GA SCH07362049GA SCH07362053GA SCH07362056GA SCH07372059GA
SCH07372086GA SCH07382337GA SCH07382338GA SCH07382341GA SCH07382343GA SCH07382344GA
SCH07392350GA SCH07392352GA SCH07392400GA SCH07392436GA SCH07392437GA JA104290036GA
JA104370161GA JA104380192GA JA104390295GA JA104390313GA JA104390314GA JA104390315GA
JA104390317GA JA104460552GA JA104470591GA JA104531093GA JA104531110GA JA104531116GA
JA104531143GA JA105021248GA JA105021253GA JA105071395GA JA105071411GA JA105141715GA
JA105141716GA JA105141717GA JA105141718GA JA105181851GA JA105202014GA JA105222072GA
JA105282322GA JA105342529GA JA105513687GA JA105523759GA JA105523763GA JA105523779GA
JA105523790GA JA105523794GA JA106033886GA JA106054039GA JA106074244GA JA106074253GA
JA106074290GA JA106094376GA JA106094424GA JA106094425GA JA106124553GA JA106134594GA
JA106134598GA JA106144669GA JA106144680GA JA106154726GA JA106164735GA JA106265212GA
JA106345518GA JA106345527GA JA106355556GA JA106365586GA JA106365593GA JA106365610GA
JA106395798GA SCH07110763GA SCH07150978GA SCH07161055GA SCH07281441GA SCH07321737GA
SCH07361991GA SCH07382241GA SCH07382269GA SCH07382292GA SCH07382310GA SCH07382315GA
SCH07382316GA SCH07382340GA SCH07392384GA SCH07321750GA SCH07341849GA SCH06430021GA
SCH06430031GA SCH06430044GA SCH06430047GA SCH06440068GA SCH06440099GA SCH06440160GA
SCH06440178GA SCH06440191GA SCH06440197GA SCH06450206GA SCH06510424GA SCH07040492GA
SCH07040495GA SCH07050514GA SCH07050527GA SCH07050537GA SCH07050558GA SCH07060580GA
SCH07060600GA SCH07070622GA SCH07070627GA SCH07070648GA SCH07110748GA SCH07110768GA
SCH07110770GA SCH07110780GA SCH07120827GA SCH07130840GA SCH07130843GA SCH07130845GA
SCH07130848GA SCH07130849GA SCH07130851GA SCH07130854GA SCH07140887GA SCH07150950GA
SCH07150964GA SCH07150999GA SCH07151004GA SCH07181100GA SCH07211200GA SCH07221240GA
SCH07281432GA SCH07281445GA SCH07291478GA SCH07301527GA SCH07301536GA SCH07301537GA
SCH07321725GA SCH07331754GA SCH07331804GA SCH07331834GA SCH07341840GA SCH07341845GA
SCH07341846GA SCH07341850GA SCH07341866GA SCH07361972GA SCH07362031GA SCH07372075GA
SCH07372085GA SCH07372167GA SCH07382202GA SCH07382245GA SCH07382270GA SCH07382271GA
SCH07382277GA SCH07382278GA SCH07382290GA SCH07382293GA SCH07382296GA SCH07382307GA
SCH07382308GA SCH07382312GA SCH07382317GA SCH07382331GA SCH07382332GA SCH07382335GA
SCH07382336GA SCH07392348GA SCH07392349GA SCH07392351GA SCH07392355GA SCH07392356GA
SCH07392358GA SCH07392361GA SCH07392362GA SCH07392364GA SCH07392372GA SCH07392375GA
SCH07392376GA SCH07392379GA SCH07392381GA SCH07392387GA SCH07392390GA SCH07392391GA
SCH07392404GA SCH07291463GA SCH07291477GA SCH07291493GA SCH07291496GA SCH07372133GA
SCH07382205GA SCH07382209GA SCH07382276GA SCH07382283GA SCH07382306GA SCH07382319GA
SCH07382323GA SCH07211225GA SCH07291485GA SCH06430032GA SCH06430046GA SCH06440075GA
SCH06440085GA SCH06440087GA SCH06440088GA SCH06440089GA SCH06440091GA SCH06440097GA
SCH06440130GA SCH06440134GA SCH06440135GA SCH06440138GA SCH06440141GA SCH06440154GA
SCH06440192GA SCH06450225GA SCH06450229GA SCH06450236GA SCH06450246GA SCH06450247GA
SCH06450251GA SCH06450253GA SCH06450275GA SCH06470359GA SCH06470360GA SCH06470362GA
SCH06470364GA SCH06510386GA SCH06510387GA SCH06510392GA SCH06510393GA SCH06510395GA
SCH06510396GA SCH06510400GA SCH06510401GA SCH06510402GA SCH06510403GA SCH06510405GA
SCH06510406GA SCH06510420GA SCH07040487GA SCH07040488GA SCH07040491GA SCH07050497GA
SCH07050515GA SCH07050532GA SCH07050533GA SCH07050549GA SCH07050557GA SCH07060583GA
SCH07060588GA SCH07060593GA SCH07060602GA SCH07070616GA SCH07070624GA SCH07070629GA
SCH07070632GA SCH07070633GA SCH07070636GA SCH07070637GA SCH07070640GA SCH07070642GA
SCH07070643GA SCH07070644GA SCH07070647GA SCH07070649GA SCH07070656GA SCH07070657GA
SCH07080684GA SCH07080723GA SCH07080736GA SCH07110753GA SCH07110755GA SCH07110759GA
SCH07110766GA SCH07110776GA SCH07110783GA SCH07110785GA SCH07120823GA SCH07120831GA
SCH07130841GA SCH07130842GA SCH07130853GA SCH07130859GA SCH07130861GA SCH07130863GA
SCH07130880GA SCH07130881GA SCH07140884GA SCH07140886GA SCH07140888GA SCH07140891GA
SCH07140893GA SCH07140908GA SCH07140914GA SCH07140916GA SCH07140919GA SCH07140932GA
SCH07140933GA SCH07150947GA SCH07150957GA SCH07150958GA SCH07150962GA SCH07150966GA
SCH07150967GA SCH07150972GA SCH07150973GA SCH07150977GA SCH07150979GA SCH07150983GA
SCH07150989GA SCH07161008GA SCH07161017GA SCH07161020GA SCH07161021GA SCH07161022GA
SCH07161026GA SCH07161027GA SCH07161031GA SCH07161035GA SCH07161060GA SCH07161064GA
SCH07161065GA SCH07161066GA SCH07161072GA SCH07161073GA SCH07171075GA SCH07171085GA
SCH07181105GA SCH07181106GA SCH07191112GA SCH07191118GA SCH07191119GA SCH07191121GA
SCH07191122GA SCH07191124GA SCH07191125GA SCH07191128GA SCH07191130GA SCH07191132GA
SCH07191136GA SCH07191137GA SCH07191138GA SCH07191139GA SCH07191140GA SCH07191157GA
SCH07201167GA SCH07201176GA SCH07201177GA SCH07211203GA SCH07211213GA SCH07211224GA
SCH07221249GA SCH07231269GA SCH07231270GA SCH07231282GA SCH07231289GA SCH07231290GA
SCH07231298GA SCH07231300GA SCH07231304GA SCH07231307GA SCH07231308GA SCH07231313GA
SCH07231317GA SCH07231318GA SCH07261332GA SCH07261352GA SCH07261366GA SCH07281416GA
SCH07281421GA SCH07301576GA SCH07301579GA SCH07321711GA SCH07321730GA SCH07321738GA
SCH07321751GA SCH07331809GA SCH07331813GA SCH07331814GA SCH07331815GA SCH07331818GA
SCH07331820GA SCH07331821GA SCH07331825GA SCH07341844GA SCH07361938GA SCH07361984GA
SCH07372057GA SCH07372077GA SCH07372111GA SCH07372173GA SCH07372175GA SCH07372176GA
SCH07372178GA SCH07382212GA SCH07382222GA SCH07382226GA SCH07382238GA SCH07382260GA
SCH07382261GA SCH07382265GA JA104380184GA JA104380270GA JA104380290GA JA104460537GA
JA104490730GA JA104500843GA JA105021220GA JA105021246GA JA105021255GA JA105021273GA
JA105121649GA JA105121658GA JA105252223GA JA105342519GA JA105402884GA JA106074259GA
JA106104445GA JA106405936GA SCH06440177GA SCH06450265GA SCH06450269GA SCH06450272GA
SCH06450273GA SCH06460278GA SCH06460281GA SCH06470319GA SCH07050512GA SCH07050540GA
SCH07060599GA SCH07110761GA SCH07150953GA SCH07161030GA SCH07291486GA SCH07311620GA
SCH07321732GA SCH07321735GA SCH07321743GA SCH07331808GA SCH07331835GA SCH07382234GA
SCH07382239GA SCH07382314GA SCH07382342GA SCH07382345GA SCH07392383GA JA105342535GA
JA106375632GA SCH07161050GA SCH07231314GA SCH07331812GA SCH07372153GA SCH07372159GA
SCH07372164GA SCH06450252GA JA105091475GA SCH07261386GA SCH07291480GA SCH06430014GA
SCH06440179GA SCH06450228GA SCH06450234GA SCH06450242GA SCH06450254GA SCH06450255GA
SCH06450256GA SCH06450257GA SCH06450258GA SCH06450259GA SCH06450260GA SCH06450267GA
SCH06450274GA SCH06460277GA SCH06460280GA SCH06470321GA SCH06470323GA SCH06470325GA
SCH06470326GA SCH06470330GA SCH06470334GA SCH06470337GA SCH06470339GA SCH07050541GA
SCH07120797GA SCH07161069GA SCH07181096GA SCH07311618GA SCH07311624GA SCH07311627GA
SCH07311628GA SCH07382191GA JA105222060GA JA106064055GA JA106064056GA JA106064059GA
JA106064060GA JA106385739GA JA106385746GA JA106385747GA JA106385750GA JA106395759GA
JA106395760GA JA106395777GA JA106395796GA JA106415995GA JA106466172GA SCH06430008GA
SCH06430011GA SCH06430012GA SCH06430013GA SCH06430015GA SCH06430016GA SCH06430020GA
SCH06430022GA SCH06430023GA SCH06430024GA SCH06430025GA SCH06430026GA SCH06430027GA
SCH06430028GA SCH06430029GA SCH06430038GA SCH06440050GA SCH06440057GA SCH06440062GA
SCH06440142GA SCH06440145GA SCH06440146GA SCH06440148GA SCH06440149GA SCH06440150GA
SCH06440152GA SCH06440153GA SCH06440157GA SCH06440158GA SCH06440159GA SCH06440163GA
SCH06440164GA SCH06440167GA SCH06440173GA SCH06440198GA SCH06440203GA SCH06440205GA
SCH06460287GA SCH07040493GA SCH07050502GA SCH07050510GA SCH07050554GA SCH07060570GA
SCH07060572GA SCH07060584GA SCH07060592GA SCH07060598GA SCH07070611GA SCH07070612GA
SCH07070620GA SCH07070641GA SCH07070652GA SCH07070655GA SCH07140903GA SCH07191162GA
SCH07211211GA SCH07211212GA SCH07261346GA SCH07261350GA SCH07261359GA SCH07301528GA
SCH07301547GA SCH07301552GA SCH07301561GA SCH07301563GA SCH07301570GA SCH07301578GA
SCH07311588GA SCH07311589GA SCH07311590GA SCH07311594GA SCH07311617GA SCH07311621GA
SCH07311625GA SCH07321691GA SCH07321703GA SCH07321723GA SCH07321731GA SCH07321736GA
SCH07321740GA SCH07321741GA SCH07321753GA SCH07331764GA SCH07331765GA SCH07331769GA
SCH07331770GA SCH07331772GA SCH07331779GA SCH07331784GA SCH07331785GA SCH07331787GA
SCH07331789GA SCH07331790GA SCH07331791GA SCH07331794GA SCH07331795GA SCH07331796GA
SCH07331800GA SCH07331801GA SCH07331802GA SCH07331807GA SCH07331816GA SCH07331823GA
SCH07331826GA SCH07331832GA SCH07351875GA SCH07351886GA SCH07351889GA SCH07361900GA
SCH07361914GA SCH07361916GA SCH07361989GA SCH07361998GA SCH07362008GA SCH07372087GA
SCH07372088GA SCH07372089GA SCH07372090GA SCH07372103GA SCH07372104GA SCH07372105GA
SCH07372108GA SCH07372118GA SCH07382330GA JA105191913GA JA105191924GA JA105191932GA
JA105191936GA JA105191947GA JA105232110GA JA105242156GA JA105242182GA JA105242184GA
JA105252186GA JA105252187GA JA105252189GA JA105252208GA JA105262317GA JA106225038GA
JA106285245GA JA106295332GA JA106295387GA JA106295394GA JA106335493GA JA106345506GA
JA106345530GA JA106345532GA JA106355543GA JA106355555GA JA106355559GA JA106355561GA
JA106355563GA JA106355564GA JA106355565GA JA106355572GA JA106365585GA JA106415990GA
JA106416003GA JA106416040GA JA106415984GA SCH07060596GA SCH07321746GA SCH07382258GA
JA105191910GA JA105212036GA JA105232106GA JA105242165GA JA105262250GA JA105262256GA
JA105322406GA JA105332419GA JA105433021GA JA105463200GA JA106124580GA JA106235062GA
SCH07140895GA JA105262285GA JA105433050GA JA105453132GA JA105453171GA JA105463188GA
SCH07120822GA SCH07130852GA SCH07130860GA SCH07130872GA SCH07130874GA SCH07150970GA
SCH07150971GA SCH07150990GA SCH07151002GA SCH07161013GA SCH07161033GA SCH07161046GA
SCH07362024GA SCH06440125GA SCH06440172GA SCH06440180GA SCH06440182GA SCH06440183GA
SCH06440201GA SCH06450216GA SCH06450220GA SCH06450222GA SCH06450226GA SCH06450237GA
SCH06470350GA SCH06470358GA SCH07050501GA SCH07050539GA SCH07080726GA SCH07080735GA
SCH07080742GA SCH07110771GA SCH07110774GA SCH07120802GA SCH07120805GA SCH07120807GA
SCH07120813GA SCH07120828GA SCH07130875GA SCH07130878GA SCH07140909GA SCH07140910GA
SCH07140911GA SCH07140920GA SCH07140921GA SCH07140923GA SCH07140927GA SCH07150944GA
SCH07191158GA SCH07201183GA SCH07261353GA SCH07261372GA SCH07261382GA SCH07261389GA
SCH07261390GA SCH07261393GA SCH07281399GA SCH07281400GA SCH07281401GA SCH07281402GA
SCH07281403GA SCH07281404GA SCH07281407GA SCH07281408GA SCH07281409GA SCH07281410GA
SCH07281411GA SCH07281431GA SCH07281440GA SCH07281442GA SCH07281447GA SCH07281449GA
SCH07281451GA SCH07281454GA SCH07281457GA SCH07291465GA SCH07291475GA SCH07291483GA
SCH07291488GA SCH07291490GA SCH07291495GA SCH07291499GA SCH07291500GA SCH07291501GA
SCH07291502GA SCH07291503GA SCH07291504GA SCH07291505GA SCH07291506GA SCH07291507GA
SCH07291509GA SCH07291510GA SCH07291512GA SCH07301515GA SCH07301520GA SCH07301521GA
SCH07301526GA SCH07301538GA SCH07301549GA SCH07301559GA SCH07311584GA SCH07311585GA
SCH07311595GA SCH07311596GA SCH07311598GA SCH07311599GA SCH07311600GA SCH07311602GA
SCH07311603GA SCH07311604GA SCH07311605GA SCH07311606GA SCH07311607GA SCH07311609GA
SCH07311610GA SCH07311612GA SCH07311622GA SCH07311626GA SCH07311629GA SCH07311630GA
SCH07311631GA SCH07311633GA SCH07311634GA SCH07311635GA SCH07311638GA SCH07311642GA
SCH07311643GA SCH07311644GA SCH07311653GA SCH07321654GA SCH07321655GA SCH07321656GA
SCH07321657GA SCH07321658GA SCH07321660GA SCH07321661GA SCH07321662GA SCH07321664GA
SCH07321665GA SCH07321666GA SCH07321668GA SCH07321669GA SCH07321670GA SCH07321671GA
SCH07321673GA SCH07321674GA SCH07321675GA SCH07321676GA SCH07321677GA SCH07321678GA
SCH07321679GA SCH07321680GA SCH07321683GA SCH07321685GA SCH07321687GA SCH07321688GA
SCH07321693GA SCH07321706GA SCH07321707GA SCH07321710GA SCH07321726GA SCH07321742GA
SCH07321749GA SCH07331788GA SCH07331829GA SCH07331836GA SCH07341858GA SCH07341859GA
SCH07341865GA SCH07341873GA SCH07351888GA SCH07351891GA SCH07351892GA SCH07361901GA
SCH07361908GA SCH07361912GA SCH07361924GA SCH07361931GA SCH07361937GA SCH07361940GA
SCH07361948GA SCH07361949GA SCH07362021GA SCH07362032GA SCH07372095GA SCH07372097GA
SCH07372100GA SCH07372102GA SCH07382231GA SCH07382232GA SCH07382233GA SCH07382242GA
SCH07382253GA SCH07382254GA SCH07382264GA SCH07382324GA SCH07382326GA JA105171813GA
JA105171836GA JA105191941GA JA105191942GA JA105191958GA JA105201984GA JA105252197GA
JA105433035GA JA106023822GA JA106023827GA JA106395888GA SCH06510408GA SCH06510435GA
SCH07110788GA SCH07110792GA SCH07130857GA SCH07130858GA SCH07140928GA SCH07140934GA
SCH07150984GA SCH07150996GA SCH07161029GA SCH07161057GA SCH07171077GA SCH07171078GA
SCH07171080GA SCH07181087GA SCH07181088GA SCH07181090GA SCH07181091GA SCH07181094GA
SCH07181095GA SCH07191108GA SCH07191126GA SCH07201194GA SCH07211214GA SCH07281430GA
SCH07311651GA SCH07321659GA SCH07321663GA SCH07321681GA SCH07321682GA SCH07321692GA
SCH07321701GA SCH07321705GA SCH07321708GA SCH07321709GA SCH07321713GA SCH07321727GA
SCH07321728GA SCH07321729GA SCH07321739GA SCH07321747GA SCH07331757GA SCH07331758GA
SCH07331761GA SCH07331763GA SCH07331767GA SCH07331768GA SCH07331773GA SCH07331775GA
SCH07331776GA SCH07331777GA SCH07331778GA SCH07331781GA SCH07331782GA SCH07331786GA
SCH07331797GA SCH07361936GA SCH07372065GA SCH07372069GA SCH07372071GA SCH07382273GA
SCH07382304GA SCH07382322GA SCH07382333GA SCH07392399GA SCH07392403GA SCH07392407GA
SCH07392410GA SCH07392419GA JA104510945GA JA104510966GA JA104520987GA JA104521005GA
JA105071405GA JA105071412GA JA106074263GA JA106094387GA JA106094392GA JA106094395GA
JA106164765GA JA106174804GA JA106184838GA JA106184847GA JA106184851GA JA106184852GA
JA106184858GA JA106184860GA JA107096744GA JA107096746GA JA107096747GA JA107096748GA
JA107096751GA JA107096752GA JA107096755GA JA107096756GA JA107096774GA SCH06430037GA
SCH07201185GA SCH07201186GA SCH07211205GA SCH07311616GA SCH07321696GA SCH07321697GA
SCH07321719GA SCH07321720GA SCH07382216GA SCH07382227GA SCH07382246GA SCH06430033GA
SCH06440059GA SCH07040473GA SCH07040478GA SCH07040482GA SCH07110746GA SCH07110765GA
SCH07150945GA SCH07261365GA SCH07261376GA SCH07261384GA SCH07261385GA SCH07351878GA
SCH07351879GA SCH07351884GA SCH07351885GA SCH07361909GA SCH07362025GA SCH07382309GA
JA106395812GA SCH07030458GA SCH07130883GA SCH07201178GA SCH07392460GA SCH06440100GA
SCH06510436GA SCH07030464GA SCH07050528GA SCH07070619GA SCH07080695GA SCH07080699GA
SCH07080709GA SCH07080716GA SCH07110750GA SCH07120829GA SCH07372091GA SCH07382279GA
SCH07382288GA SCH07382291GA SCH07382302GA SCH07392392GA SCH07392395GA SCH07392396GA
SCH07392397GA SCH07392406GA SCH07392409GA JA105021203GA JA106305432GA JA106305450GA
JA106516571GA JA106516573GA JA106516578GA SCH07040475GA SCH07150997GA SCH07311652GA
SCH07321684GA SCH07321712GA SCH07321744GA JA105483353GA SCH06230001GA SCH06230002GA
SCH06320005GA SCH06320006GA SCH07050526GA SCH07050531GA SCH07050547GA SCH07050550GA
SCH07050551GA SCH07070614GA SCH07150992GA SCH07221252GA SCH07221261GA SCH07231310GA
SCH07261397GA SCH07301546GA SCH07331805GA SCH07331824GA SCH07341843GA SCH07361915GA
SCH07361946GA SCH07361950GA SCH07362045GA SCH07362050GA SCH07362051GA SCH07372058GA
SCH07372064GA SCH07372099GA SCH07372117GA SCH07382200GA SCH07382240GA SCH07382243GA
SCH07382329GA SCH07392347GA SCH07392354GA JA105463225GA JA105463228GA JA106023833GA
JA106023851GA JA106486197GA JA106486210GA JA106486211GA JA105191959GA JA106285243GA
SCH07060571GA SCH07120832GA SCH07080701GA SCH07080710GA SCH07110751GA SCH07110754GA
SCH07120811GA SCH07120836GA SCH07130856GA SCH07130862GA SCH07130864GA SCH07130868GA
SCH07130869GA SCH07281405GA SCH07281425GA SCH07321722GA SCH07341838GA SCH07351876GA
SCH07351890GA SCH07361893GA SCH07361902GA SCH07361911GA SCH07361926GA SCH07361932GA
SCH07361939GA SCH07372066GA SCH06440147GA SCH06440156GA SCH06440165GA SCH06440166GA
SCH06440170GA SCH06440171GA SCH06440174GA SCH06440181GA SCH07040470GA SCH07050544GA
SCH07070625GA SCH07080674GA SCH07130865GA SCH07140900GA SCH07161039GA SCH07161043GA
SCH07161044GA SCH07161045GA SCH07161047GA SCH07161051GA SCH07171079GA SCH07171083GA
SCH07181089GA SCH07191111GA SCH07191114GA SCH07191116GA SCH07201184GA SCH07261387GA
SCH07291460GA SCH07301533GA SCH07301557GA SCH07321690GA SCH07362042GA SCH07362044GA
SCH07372068GA JA105392804GA SCH06460279GA SCH06460285GA SCH06470335GA SCH06510411GA
SCH07050505GA SCH07140907GA SCH07140917GA SCH07140935GA SCH07140940GA SCH07231301GA
SCH07372062GA SCH07372101GA SCH07382199GA SCH07382204GA SCH07382213GA SCH07382219GA
SCH07382220GA SCH07382221GA JA104360105GA JA104430389GA JA105051303GA JA105101506GA
JA105101514GA JA105161762GA JA105453183GA JA106285312GA JA106345517GA JA106345528GA
JA106355568GA JA106355576GA JA106365588GA JA106365595GA JA106375613GA JA106375614GA
JA106375615GA JA106375617GA JA106375618GA JA106395856GA JA106395859GA JA106395870GA
JA106395871GA JA106395872GA JA106395873GA JA106395874GA JA106395875GA JA106395879GA
JA106395884GA JA106395891GA JA106395897GA JA106405935GA JA106405968GA JA106405969GA
JA106416010GA JA106486187GA JA106486250GA JA106486253GA JA106094355GA SCH06440143GA
SCH06440168GA SCH06440176GA SCH07110782GA SCH07161053GA SCH07171076GA SCH07171081GA
SCH07171084GA SCH07171086GA SCH07181093GA SCH07181097GA SCH07191109GA SCH07211220GA
SCH07331755GA SCH07341851GA SCH07361943GA SCH07382251GA SCH07382256GA SCH07382257GA
SCH07382268GA SCH07382286GA SCH07382300GA SCH07392357GA SCH07392388GA JA104380218GA
JA104390305GA JA104440464GA JA106164760GA JA106204942GA JA106204945GA JA106204948GA
JA106204955GA JA106214988GA JA106214992GA JA106214998GA JA106215000GA JA106215014GA
JA106215016GA JA106215021GA JA106215023GA JA106225025GA JA106225053GA JA106235065GA
JA106235069GA JA106235070GA JA106235071GA JA106235072GA JA106375624GA JA106375636GA
JA106375641GA JA107026678GA JA107036702GA JA107036704GA JA107036707GA JA107036709GA
JA107036710GA JA107036717GA JA107036718GA JA107036719GA JA107036722GA JA107036726GA
JA107036729GA SCH07030466GA SCH07050543GA SCH07090745GA SCH07040479GA SCH07050499GA
SCH07060579GA SCH07060586GA SCH07070630GA SCH07070635GA SCH07080740GA SCH07110772GA
SCH07120810GA SCH07150988GA SCH07161034GA SCH07191141GA SCH07191145GA SCH07191146GA
SCH07191147GA SCH07191155GA SCH07191164GA SCH07201168GA SCH07231292GA SCH07231309GA
SCH07281406GA SCH07281414GA SCH07281415GA SCH07281424GA SCH07281433GA SCH07281437GA
SCH07311619GA SCH07321700GA SCH07331793GA SCH07331806GA SCH07331810GA SCH07331811GA
SCH07331817GA SCH07331822GA SCH07361899GA SCH07361903GA SCH07361907GA SCH07361913GA
SCH07361921GA SCH07361922GA SCH07361925GA SCH07382294GA SCH07382325GA SCH07392411GA
SCH07392420GA JA105513611GA JA105513614GA JA106285264GA SCH07392461GA SCH07040494GA
SCH07080717GA SCH07080727GA SCH07080729GA SCH07110760GA SCH07110779GA SCH07130873GA
SCH07140905GA SCH07140913GA SCH07140915GA SCH07140926GA SCH07140929GA SCH07372070GA
SCH07382259GA SCH07382295GA SCH07382313GA SCH07382321GA SCH07392373GA SCH07392380GA
SCH07392385GA SCH07392386GA SCH07392417GA JA105402895GA SCH06430017GA SCH06430019GA
SCH07050504GA SCH07060563GA SCH07070623GA SCH07080721GA SCH07120814GA SCH07130846GA
SCH07161067GA SCH07161074GA SCH07171082GA SCH07191110GA SCH07211223GA SCH07211226GA
SCH07231295GA SCH07261340GA SCH07372092GA SCH07060564GA SCH07070653GA SCH07191115GA
SCH07211198GA SCH07211202GA SCH07211204GA SCH07281422GA SCH07331799GA SCH07211201GA
SCH07211206GA SCH07211209GA SCH07221230GA SCH07221242GA SCH07321672GA SCH07321714GA
SCH07361917GA SCH07361918GA SCH07361919GA SCH07361920GA SCH07361923GA SCH07361933GA
SCH07361941GA SCH07361942GA JA104500798GA JA104500801GA JA104500804GA JA104500805GA
SCH07110749GA SCH07261368GA SCH07311614GA SCH06470329GA SCH06470341GA SCH07231276GA
SCH07231285GA SCH06450270GA SCH06460305GA SCH06470369GA SCH07050513GA SCH07050552GA
SCH07060589GA SCH07070617GA SCH07070628GA SCH07070631GA SCH07070659GA SCH07070664GA
SCH07080670GA SCH07080681GA SCH07181103GA SCH07321745GA SCH07321748GA SCH07331774GA
SCH07331792GA SCH07351883GA SCH07351887GA SCH07361894GA SCH07361929GA SCH07361990GA
SCH07362022GA JA106285289GA JA106285315GA JA106295318GA JA106295319GA JA106295320GA
JA106295322GA JA106295323GA JA106295324GA JA106295334GA JA106295335GA JA106295336GA
JA106295337GA JA106295347GA SCH06440061GA SCH07050507GA SCH07050536GA SCH07060601GA
SCH07070607GA SCH07120796GA SCH07120835GA SCH07150986GA SCH07161010GA SCH07161059GA
SCH07161061GA SCH07191163GA SCH07201170GA SCH07201181GA SCH07201189GA SCH07201191GA
SCH07211207GA SCH07211210GA SCH07211215GA SCH07211216GA SCH07211218GA SCH07211219GA
SCH07211221GA SCH07221237GA SCH07221241GA SCH07221245GA SCH07231316GA SCH07261333GA
SCH07261335GA SCH07261336GA SCH07261338GA SCH07261339GA SCH07261347GA SCH07261349GA
SCH07261351GA SCH07261354GA SCH07261358GA SCH07261363GA SCH07261364GA SCH07261373GA
SCH07261379GA SCH07261388GA SCH07281418GA SCH07281423GA SCH07281426GA SCH07291459GA
SCH07361983GA SCH07372063GA SCH07372074GA SCH07372149GA SCH07372172GA SCH07372174GA
SCH07372180GA SCH07372181GA SCH07372183GA SCH07382187GA SCH07382190GA SCH07382193GA
SCH07382218GA SCH07382280GA SCH07382287GA SCH07392360GA SCH07392365GA SCH07392366GA
SCH07392367GA SCH07392368GA SCH07392369GA SCH07392370GA SCH07392371GA SCH07392374GA
SCH07392377GA SCH07392382GA SCH07392412GA JA104470649GA JA104480665GA JA104480686GA
JA105021274GA JA106204924GA JA106235086GA SCH07191127GA SCH07191152GA SCH07060577GA
SCH07080694GA SCH07221250GA SCH07221255GA SCH07221258GA SCH07221259GA SCH07231264GA
SCH07291469GA SCH07301530GA SCH07341842GA SCH07341848GA SCH07341863GA SCH07341868GA
SCH07361970GA SCH07372123GA SCH07372151GA JA105322360GA JA105322371GA JA105513643GA
JA105513681GA JA105513696GA SCH06440169GA SCH06470355GA SCH06510407GA SCH07050508GA
SCH07050511GA SCH07050517GA SCH07070663GA SCH07070665GA SCH07080677GA SCH07151001GA
SCH07161037GA SCH07161070GA SCH07181099GA SCH07261331GA SCH07291491GA SCH07392393GA
SCH07392394GA SCH07392408GA JA104531112GA JA104531114GA JA104531117GA JA105483270GA
JA106043927GA JA106134627GA JA106134629GA JA106315454GA SCH06440140GA SCH06510399GA
SCH06510432GA SCH07030453GA SCH07030455GA SCH07030463GA SCH07191165GA SCH07201166GA
SCH07201182GA SCH07231288GA SCH07361965GA SCH07361966GA SCH07361971GA SCH07392429GA
SCH07392430GA SCH07392444GA SCH06440098GA SCH06440105GA SCH06440127GA SCH06450261GA
SCH06510394GA SCH07040476GA SCH07040483GA SCH07040484GA SCH07050509GA SCH07050519GA
SCH07050553GA SCH07070621GA SCH07070638GA SCH07070645GA SCH07080673GA SCH07080702GA
SCH07080703GA SCH07080718GA SCH07080730GA SCH07080737GA SCH07080744GA SCH07110767GA
SCH07110786GA SCH07120793GA SCH07120801GA SCH07120804GA SCH07120808GA SCH07120815GA
SCH07120824GA SCH07150969GA SCH07150975GA SCH07150976GA SCH07150980GA SCH07150985GA
SCH07150987GA SCH07150991GA SCH07150993GA SCH07150995GA SCH07150998GA SCH07161024GA
SCH07161068GA SCH07191113GA SCH07291461GA SCH07291494GA SCH07291498GA SCH07291508GA
SCH07341864GA SCH07372073GA SCH07372080GA SCH07372084GA SCH07372096GA SCH07382215GA
SCH07382252GA SCH07382263GA SCH07382285GA SCH07382297GA SCH07382305GA JA104380206GA
JA105332453GA JA105342550GA JA105352552GA JA105433010GA JA106265209GA JA106496370GA
JA106496373GA JA106496374GA JA106496381GA JA106496384GA JA106496385GA JA106496388GA
JA106496390GA JA106496401GA JA106496403GA JA106496404GA JA106496405GA JA106496408GA
JA106506415GA JA106506417GA JA106506468GA JA106506470GA JA106506471GA JA106506472GA
JA106506473GA JA106506474GA JA106506487GA JA106506512GA JA106516525GA JA106516528GA
SCH06440108GA SCH06440129GA SCH06440131GA SCH06450213GA SCH06450221GA SCH06510434GA
SCH07030459GA SCH07140939GA SCH07291474GA SCH07311636GA SCH07311637GA SCH07341870GA
SCH07361959GA SCH07372136GA SCH07372137GA SCH07372143GA SCH07372147GA SCH07372148GA
SCH07372163GA SCH07372170GA JA105493431GA SCH06460282GA SCH06460288GA SCH06460289GA
SCH06470331GA SCH06470365GA SCH06470370GA SCH07060591GA SCH07070662GA SCH07140941GA
SCH07321695GA SCH07050538GA SCH07080728GA SCH07080738GA SCH07080739GA SCH07110762GA
SCH07110778GA SCH07110784GA SCH07261356GA SCH07382223GA SCH07382244GA SCH07382248GA
JA105443088GA JA105453123GA JA105483333GA JA105483336GA JA105483342GA JA105483347GA
JA105483351GA JA105483352GA JA105483354GA JA105483358GA JA105483361GA JA105483362GA
JA105483364GA JA105493367GA JA105493369GA JA105493372GA JA105493375GA JA105493377GA
JA105493379GA JA105493384GA JA105493385GA JA105493386GA JA105493390GA JA105493391GA
JA105493396GA JA105493397GA JA105493400GA JA105493406GA JA105493407GA JA105493408GA
JA105493409GA JA105493410GA JA105493413GA JA105493416GA JA105493417GA JA105493418GA
JA105493419GA JA105493420GA JA105493424GA JA105493425GA JA105493427GA JA105493429GA
JA105493432GA JA105493433GA JA105493434GA JA105493435GA JA105493436GA JA105493437GA
JA105493440GA JA105493441GA JA105493445GA JA105493446GA JA105493447GA JA105493448GA
JA105493449GA JA105493450GA JA105493451GA JA105493453GA JA105493455GA JA105503461GA
JA105503462GA JA105503465GA JA105503470GA JA105503472GA JA105503473GA JA106013815GA
JA106023848GA JA106104479GA JA106204962GA SCH06440093GA SCH06440101GA SCH06440102GA
SCH06440107GA SCH06460300GA SCH06460309GA SCH07070660GA SCH07201193GA SCH07231293GA
SCH07060565GA SCH06460302GA SCH06460310GA SCH06460312GA SCH07050518GA SCH07080683GA
SCH07040471GA SCH07070654GA SCH07070661GA SCH07080671GA SCH07120794GA SCH07161023GA
SCH07191142GA SCH07191143GA SCH07191144GA SCH07191150GA SCH07191151GA SCH07191153GA
SCH07191159GA SCH07301517GA JA105161760GA JA106124571GA JA106204967GA SCH06440081GA
SCH06470327GA SCH06470354GA SCH06470361GA SCH07050525GA SCH07050560GA SCH07060567GA
SCH07060595GA SCH07070608GA SCH07080719GA SCH07110787GA SCH07221236GA SCH07221238GA
SCH07221239GA SCH07221243GA SCH07221247GA SCH07221253GA SCH07231263GA SCH07361986GA
SCH07362011GA SCH07382217GA SCH07382235GA SCH07382236GA SCH07382237GA SCH07382303GA
SCH07382334GA SCH06440104GA SCH06470316GA SCH07060569GA SCH07150959GA SCH07150965GA
SCH07161018GA SCH07161032GA SCH07221257GA SCH07221260GA SCH07231262GA SCH07231265GA
SCH07231267GA SCH07231274GA SCH07231275GA SCH07231277GA SCH07231278GA SCH07231279GA
SCH07372157GA SCH06430010GA SCH06440070GA SCH06440071GA SCH06440078GA SCH06440086GA
SCH06440090GA SCH06440092GA SCH06440096GA SCH06440106GA SCH06440109GA SCH06440110GA
SCH06440111GA SCH06440113GA SCH06440115GA SCH06440116GA SCH06440124GA SCH06440132GA
SCH06440137GA SCH06440139GA SCH06440199GA SCH06510397GA SCH06510416GA SCH06510422GA
SCH06510423GA SCH06510425GA SCH07040489GA SCH07040490GA SCH07050523GA SCH07080689GA
SCH07080696GA SCH07080704GA SCH07080732GA SCH07080734GA SCH07080741GA SCH07110752GA
SCH07110758GA SCH07110769GA SCH07110773GA SCH07110775GA SCH07110781GA SCH07110789GA
SCH07110791GA SCH07120795GA SCH07120798GA SCH07120799GA SCH07120800GA SCH07120803GA
SCH07120812GA SCH07120817GA SCH07120819GA SCH07120820GA SCH07130866GA SCH07140885GA
SCH07140918GA SCH07140925GA SCH07150946GA SCH07150968GA SCH07150974GA SCH07161011GA
SCH07161028GA SCH07161071GA SCH07181098GA SCH07191120GA SCH07191161GA SCH07201172GA
SCH07201192GA SCH07201195GA SCH07211208GA SCH07221248GA SCH07231315GA SCH07261319GA
SCH07261320GA SCH07261321GA SCH07261322GA SCH07261323GA SCH07261324GA SCH07261325GA
SCH07261326GA SCH07261327GA SCH07261328GA SCH07261329GA SCH07261330GA SCH07261337GA
SCH07261344GA SCH07261345GA SCH07261348GA SCH07261355GA SCH07261357GA SCH07261360GA
SCH07261361GA SCH07261374GA SCH07261392GA SCH07261396GA SCH07281412GA SCH07281413GA
SCH07281417GA SCH07281419GA SCH07281420GA SCH07281427GA SCH07281428GA SCH07281435GA
SCH07281436GA SCH07281438GA SCH07281439GA SCH07281443GA SCH07281444GA SCH07291472GA
SCH07291473GA SCH07291489GA SCH07291497GA SCH07311601GA SCH07321733GA SCH07321734GA
SCH07331798GA SCH07351877GA SCH07361895GA SCH07361896GA SCH07361898GA SCH07361928GA
SCH07362007GA SCH07362043GA SCH07372060GA SCH07372061GA SCH07372076GA SCH07372079GA
SCH07372083GA SCH07372098GA SCH07372106GA SCH07372107GA SCH07372109GA SCH07372110GA
SCH07372112GA SCH07372113GA SCH07372115GA SCH07372116GA SCH07372184GA SCH07372185GA
SCH07372186GA SCH07382188GA SCH07382189GA SCH07382192GA SCH07382194GA SCH07382195GA
SCH07382197GA SCH07382198GA SCH07382201GA SCH07382203GA SCH07382206GA SCH07382207GA
SCH07382208GA SCH07382210GA SCH07382211GA SCH07382249GA SCH07382299GA SCH07392418GA
SCH07392423GA SCH07392425GA JA104440445GA JA104500837GA JA105201999GA JA105202015GA
JA105202016GA JA105202017GA JA106385743GA JA106395771GA JA106395775GA SCH07201175GA
JA107026659GA SCH07080712GA SCH07201169GA SCH07361935GA SCH06470342GA SCH07050522GA
SCH07140937GA SCH07150981GA SCH07361904GA SCH07130837GA SCH07130847GA SCH07341867GA
SCH06430034GA SCH06430042GA SCH06440054GA SCH06440079GA SCH06440112GA SCH06440118GA
SCH06440120GA SCH06450207GA SCH06450240GA SCH06450243GA SCH06450248GA SCH06450249GA
SCH06450250GA SCH06450266GA SCH06450268GA SCH06460286GA SCH06470332GA SCH06510385GA
SCH06510413GA SCH07050545GA SCH07080711GA SCH07130876GA SCH07140930GA SCH07140931GA
SCH07140943GA SCH07150949GA SCH07150954GA SCH07181104GA SCH07181107GA SCH07191131GA
SCH07191134GA SCH07211222GA SCH07221231GA SCH07221232GA SCH07221246GA SCH07261342GA
SCH07331803GA SCH07372179GA SCH07372182GA SCH07382262GA SCH07382320GA SCH07392363GA
SCH07392413GA SCH07392415GA SCH07392416GA SCH07392422GA SCH07392438GA SCH07392439GA
SCH07392441GA SCH07392445GA SCH07392453GA SCH07392456GA SCH07392467GA SCH07392468GA
SCH06450224GA SCH06460304GA SCH06470347GA SCH07030460GA SCH07040477GA SCH07040481GA
SCH07040485GA SCH07070613GA SCH07130855GA SCH07181101GA SCH07231281GA SCH07261334GA
SCH07261367GA SCH07261370GA SCH07261375GA SCH07261378GA SCH07261380GA SCH07261394GA
SCH07281434GA SCH07281450GA SCH07281453GA SCH07281455GA SCH07331759GA SCH07331830GA
SCH07331837GA SCH07351874GA SCH07361973GA SCH07050548GA SCH07382196GA SCH07382214GA
SCH07382272GA SCH07392378GA SCH07040486GA SCH07050521GA SCH07050535GA SCH07050542GA
SCH07110790GA SCH07130877GA SCH07130882GA SCH07151005GA SCH07191135GA SCH07321752GA
SCH07331756GA SCH07361930GA SCH07361944GA SCH07361945GA SCH07361947GA SCH07361951GA
SCH07361952GA SCH07361953GA SCH07361954GA SCH07361955GA SCH07361961GA SCH07361962GA
SCH07361963GA SCH07361979GA SCH07361980GA SCH07361981GA SCH07361982GA SCH07361985GA
SCH07361988GA SCH07361992GA SCH07361993GA SCH07361994GA SCH07361995GA SCH07361996GA
SCH07361997GA SCH07361999GA SCH07362000GA SCH07362001GA SCH07362005GA SCH07362009GA
SCH07362010GA SCH07362012GA SCH07362013GA SCH07362015GA SCH07362016GA SCH07362017GA
SCH07362018GA SCH07362019GA SCH07362020GA SCH07362030GA SCH07362038GA SCH07362039GA
SCH07362041GA SCH07362052GA SCH07362054GA SCH07362055GA SCH07372072GA SCH07372093GA
SCH07372094GA SCH07392405GA JA104380228GA JA104380248GA JA104380266GA JA104521002GA
JA105513557GA JA106144662GA JA106174806GA SCH07080715GA SCH07161058GA JA105232089GA
JA104440467GA JA106506477GA JA104430398GA JA105402885GA JA105433013GA JA105433019GA
JA105433024GA JA105151747GA JA106154714GA JA106164758GA JA106355575GA JA106365583GA
JA105473247GA JA105473248GA JA105513622GA JA105523710GA JA106215024GA JA106225041GA
JA106225044GA JA106064150GA JA106385658GA JA106204933GA JA106214991GA SCH07040474GA
SCH07050503GA SCH07060566GA SCH07060585GA SCH07070650GA SCH07070651GA SCH07080672GA
SCH07080682GA SCH07080685GA SCH07080688GA SCH07080691GA SCH07080697GA SCH07080698GA
SCH07080700GA SCH07080705GA SCH07080706GA SCH07080725GA SCH07110747GA SCH07120833GA
SCH07140892GA SCH07140897GA SCH07140901GA SCH07140902GA SCH07231294GA SCH07231312GA
SCH07331783GA SCH07341869GA SCH07341871GA SCH07341872GA SCH07361910GA SCH07361960GA
SCH07361967GA SCH07361968GA SCH07372122GA SCH07372129GA SCH07321716GA SCH07331762GA
SCH06460284GA SCH07070639GA SCH07080733GA SCH07231283GA SCH07341855GA SCH07362034GA
SCH07331760GA SCH07331771GA SCH07331780GA SCH06440193GA SCH06440200GA SCH07140924GA
SCH07161016GA SCH07211217GA SCH07291514GA SCH07301539GA SCH07341862GA SCH07362023GA
SCH07362029GA SCH07362035GA SCH06440194GA SCH06450233GA SCH06450263GA SCH06460291GA
SCH06460303GA SCH06460307GA SCH06470313GA SCH06470314GA SCH06470345GA SCH06510404GA
SCH07030454GA SCH07030465GA SCH07030467GA SCH07030468GA SCH07050498GA SCH07050524GA
SCH07060568GA SCH07060575GA SCH07080690GA SCH07110757GA SCH07120809GA SCH07150952GA
SCH07161012GA SCH07161014GA SCH07161036GA SCH07221256GA SCH07261377GA SCH07261381GA
SCH07261395GA SCH07291462GA SCH07291464GA SCH07291466GA SCH07291470GA SCH07291484GA
SCH07361905GA SCH07361906GA SCH07372124GA SCH07281446GA SCH06450244GA SCH06460292GA
SCH06510383GA SCH07080686GA SCH07281448GA SCH07291481GA SCH07080680GA SCH07080743GA
SCH07120816GA SCH07191149GA SCH07191160GA SCH07201179GA SCH07201180GA SCH07211228GA
SCH07281452GA SCH07281456GA SCH07311645GA SCH07311649GA SCH07311650GA SCH07331819GA
SCH07331828GA SCH07351880GA SCH07351882GA SCH07361978GA SCH07362003GA SCH07362004GA
SCH07231306GA SCH07311639GA SCH07080708GA SCH06460301GA SCH06460311GA SCH07211229GA
SCH07382230GA SCH07221234GA SCH06440066GA SCH06440188GA SCH06440189GA SCH06440195GA
SCH06460283GA SCH07191129GA SCH07191133GA SCH07321689GA SCH07361974GA SCH07392440GA
SCH07392450GA SCH06470315GA SCH07070634GA SCH07070666GA SCH07130879GA SCH07181102GA
SCH07050500GA SCH07050530GA SCH07050555GA SCH07060562GA SCH07060576GA SCH07060582GA
SCH07070603GA SCH07070604GA SCH07070605GA SCH07070606GA SCH07070609GA SCH07201188GA
SCH07221251GA SCH07231271GA SCH07361958GA SCH07361969GA SCH07361976GA SCH07362040GA
SCH07372119GA SCH07372126GA SCH07372127GA SCH07382228GA SCH07382229GA SCH07070667GA
SCH06440114GA SCH07392432GA SCH07392442GA SCH07130838GA SCH07050559GA SCH07130867GA
SCH07392465GA SCH07050529GA SCH07231273GA SCH07060597GA SCH06470346GA SCH07080675GA
SCH07080676GA SCH07080678GA SCH07080679GA SCH07372144GA SCH07372158GA SCH07372160GA
SCH07372165GA SCH07372166GA SCH07372171GA SCH07201196GA SCH07201197GA SCH07331831GA
SCH07331833GA SCH07362033GA SCH07372067GA SCH07372125GA SCH07372128GA SCH06430040GA
SCH07130850GA SCH07191156GA SCH07231299GA SCH07301518GA SCH07231287GA SCH07050546GA
SCH07060561GA SCH07080692GA SCH07191117GA SCH07191123GA SCH07281458GA SCH07372161GA
JA106255173GA SCH06440095GA SCH06440103GA SCH06440186GA SCH06450218GA SCH06450276GA
SCH06460306GA SCH07040472GA SCH07070615GA SCH07070618GA SCH07070646GA SCH07080731GA
SCH07130870GA SCH07140899GA SCH07150951GA SCH07150955GA SCH07150956GA SCH07150961GA
SCH07161038GA SCH07161049GA SCH07161052GA SCH07181092GA SCH07201187GA SCH07201190GA
SCH07211199GA SCH07221235GA SCH07231286GA SCH07231305GA SCH07261362GA SCH07261398GA
SCH07291476GA SCH07291482GA SCH07341841GA SCH07341847GA SCH07351881GA SCH07361975GA
SCH07361977GA SCH07361987GA SCH07362002GA SCH07362036GA SCH07392427GA SCH07392449GA
SCH06430018GA SCH06430043GA SCH06440055GA SCH06440133GA SCH06440196GA SCH06440204GA
SCH06450211GA SCH06450212GA SCH06450239GA SCH06450245GA SCH06450271GA SCH06460290GA
SCH06470363GA SCH06510415GA SCH06510433GA SCH07030457GA SCH07040480GA SCH07060578GA
SCH07060581GA SCH07060587GA SCH07070658GA SCH07080668GA SCH07080687GA SCH07080720GA
SCH07080724GA SCH07120821GA SCH07150982GA SCH07161015GA SCH07161041GA SCH07161048GA
SCH07231266GA SCH07231291GA SCH07261341GA SCH07261369GA SCH07261371GA SCH07261383GA
SCH07281429GA SCH07291467GA SCH07372078GA SCH07392428GA SCH07392431GA JA104370151GA
JA105021186GA JA105101522GA JA105322353GA JA105322369GA JA105443062GA JA105443063GA
JA106043988GA SCH06450230GA SCH06450235GA SCH06450262GA SCH06450264GA SCH07361956GA
SCH07361957GA SCH07362037GA JA106054021GA JA106054022GA JA106054024GA JA106054025GA
JA106285262GA JA106285292GA JA106355536GA SCH06470328GA SCH07060594GA SCH07150963GA
SCH07291468GA SCH07321718GA SCH07361964GA SCH07362006GA SCH07321715GA SCH07321717GA
SCH07120818GA SCH07120826GA SCH07120834GA SCH07130839GA SCH07130844GA SCH07161056GA
SCH07201171GA SCH07201173GA SCH07201174GA SCH07211227GA SCH07221254GA SCH07231280GA
SCH07231284GA SCH07231296GA SCH07231297GA SCH07231302GA SCH07231303GA SCH07231311GA
SCH07291487GA SCH07311641GA SCH07311646GA SCH07311647GA SCH07321702GA SCH07321704GA
SCH07341839GA SCH07341854GA SCH07341856GA SCH07341857GA SCH07341860GA SCH07341861GA
SCH07372121GA SCH07372131GA SCH07372132GA SCH07372135GA SCH07372139GA SCH07372140GA
SCH07372141GA SCH07372142GA SCH07372145GA SCH07372146GA SCH07372152GA SCH07372154GA
SCH07372155GA SCH07372156GA SCH07372162GA SCH07372168GA SCH07372169GA SCH07372177GA
SCH07382224GA SCH07382225GA SCH07382328GA SCH07030461GA SCH07030462GA SCH07321724GA
SCH06440073GA SCH06440074GA SCH06470317GA SCH06470318GA SCH06470320GA SCH06470322GA
SCH06470324GA SCH06470333GA SCH06470336GA SCH06470338GA SCH06470340GA SCH06470343GA
SCH06470344GA SCH06470348GA SCH06470349GA SCH06470351GA SCH06470353GA SCH06470356GA
SCH06470357GA SCH06510384GA SCH07030450GA SCH07030451GA SCH07030452GA SCH07030456GA
SCH07040469GA SCH07050516GA SCH07060573GA SCH07060574GA SCH07080669GA SCH07080707GA
SCH07130871GA SCH07140896GA SCH07140898GA SCH07151007GA SCH07161040GA SCH07161042GA
SCH07231268GA SCH07231272GA SCH07261391GA SCH07291471GA SCH07311615GA SCH07362026GA
SCH07362028GA SCH07372081GA SCH07372082GA SCH07372134GA SCH07291479GA SCH06450231GA
SCH06450241GA SCH06460308GA SCH07080693GA SCH07080713GA SCH07080722GA SCH07110756GA
SCH07140889GA SCH07372114GA SCH07372120GA SCH06440162GA SCH06450238GA SCH07050534GA
SCH07191148GA SCH07311623GA SCH07311632GA SCH07311640GA SCH07311648GA JA107026656GA
JA107026685GA JA107026690GA JA107036711GA JA107036732GA SCH06470366GA SCH06470367GA
SCH06470368GA SCH06490371GA SCH06490372GA SCH06490373GA SCH06490374GA SCH06490375GA
SCH06490376GA SCH06520437GA SCH06520438GA SCH06520439GA SCH06520440GA SCH06520441GA
SCH06520442GA SCH06520443GA SCH06520444GA SCH06520445GA SCH06520446GA SCH06520447GA
SCH06520448GA SCH07120825GA SCH07191154GA SCH07331827GA SCH07341853GA SCH07361934GA
SCH07372130GA SCH07372138GA SCH07372150GA SCH07382247GA SCH07382250GA SCH07382255GA
SCH07382266GA SCH07382267GA SCH07382274GA SCH07382275GA SCH07382281GA SCH07382282GA
SCH07382284GA SCH07382289GA SCH07382298GA SCH07382301GA SCH07382311GA SCH07382318GA
SCH07382327GA SCH07382339GA SCH07392346GA SCH07392359GA SCH07392389GA SCH07392398GA
SCH07392401GA SCH07392402GA SCH07392414GA SCH07392421GA SCH07392424GA SCH07392426GA
SCH07392433GA SCH07392434GA SCH07392435GA SCH07392443GA SCH07392446GA SCH07392447GA
SCH07392448GA SCH07392451GA SCH07392452GA SCH07392454GA SCH07392455GA SCH07392457GA
SCH07392458GA SCH07392459GA SCH07392462GA SCH07392463GA SCH07392464GA SCH07392466GA
SCH07392469GA SCH07402470GA SCH07402471GA SCH07402472GA SCH07402473GA SCH07402474GA
SCH07402475GA SCH07402476GA SCH07402477GA SCH07402478GA SCH07402479GA SCH07402480GA
SCH07402481GA SCH07402482GA SCH07402483GA SCH07402484GA SCH07402485GA SCH07402486GA
SCH07402487GA SCH07402488GA SCH07402489GA SCH07402490GA SCH07402491GA SCH07402492GA
SCH07402493GA SCH07402494GA SCH07402495GA SCH07402496GA SCH07402497GA SCH07402498GA
SCH07402499GA SCH07402500GA SCH07402501GA SCH07402502GA SCH07402503GA SCH07402504GA
SCH07402505GA SCH07402506GA SCH07402507GA SCH07402508GA SCH07402509GA SCH07402510GA
SCH07402511GA SCH07402512GA SCH07402513GA SCH07402514GA SCH07402515GA SCH07402516GA
SCH07402517GA SCH07402518GA SCH07402519GA SCH07402520GA SCH07402521GA SCH07402522GA
SCH07402523GA SCH07402524GA SCH07402525GA SCH07402526GA SCH07402527GA SCH07402528GA
SCH07402529GA SCH07402530GA SCH07402531GA SCH07402532GA SCH07402533GA SCH07402534GA
SCH07402535GA SCH07402536GA SCH07402537GA SCH07412538GA SCH07412539GA SCH07412540GA
SCH07412541GA SCH07412542GA SCH07412543GA SCH07412544GA SCH07412545GA SCH07412546GA
SCH07412547GA SCH07412548GA SCH07412549GA SCH07412550GA SCH07412551GA SCH07412552GA
SCH07412553GA
RECALLING FIRM/MANUFACTURER
Recalling Firm: General Electric Medical Systems Information Technology, Wauwatosa,
WI, by letter on October 16, 2007.
Manufacturer: General Electric Medical Systems Information Technology, Milwaukee,
WI. Firm initiated recall is ongoing.
REASON
Potential for Loss of Audible Alarming and ECG Tab Setting Changes; from Clinical
Information Center (CIC) Pro v5.0.3 & 5.0.6 when used in combination with
Telemetry and bedside monitors.
VOLUME OF PRODUCT IN COMMERCE
3,115 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
a) Sechrist Hyperbaric Chamber Gurney, Standard Gurney & Stretcher Model,
P/N: 21502, Recall # Z-0733-2008;
b) Sechrist Hyperbaric Chamber Gurney, Standard Gurney & Deluxe Wide
Stretcher Model, P/N 21464, Recall # Z-0734-2008;
c) Sechrist Hyperbaric Chamber Gurney, Hydraulic Gurney & Deluxe Wide
Stretcher Model, P/N 21465, Recall # Z-0735-2008;
d) Sechrist Hyperbaric Chamber Gurney, Hydraulic Gurney & Standard Stretcher
Model, P/N 21466, Recall # Z-0736-2008;
e) Sechrist Hyperbaric Chamber Gurney, Standard Gurney without Side rails &
Stretcher Model, P/N 21487, Recall # Z-0737-2008
CODE
Serial Numbers: (all Model Numbers-21502, 21464, 21465, 21466. and 21487) 0604081092
0606087510 0503071536 0503071538 0408061454 0408061455 0408060131 0408061453
0609082336 0604081093 0606087506 0508081371 0508081372 0609082331 0609082333
0609079892 0506069872 0506069875 0601082091 0601082092 0607081892 0607081893
0512070332 0512070337 0604081037 0604081039 0512070333 0601083589 0502072058
0502072219 0504070085 0601083584 0609079893 0502072070 0404059009 0405060916
0406062857 0407058697 0511075672 0610079808 0612087522 0610079809 0505075450
0506070797 0506070799 0508081000 0509074722 0612087681 0612087685 0508080999
0509074952 0408066081 0411068040 0403059871 0403059875 0511075668 0511075670
0505075448 0506069874 0508081375 0508081376 0403060956 0312058058 0312058062
0309052388 0601082095 0509074723 0509074725 0402059070 0402060431 0402060433
0402060435 0502072217 0502072220 0405059703 0503071540 0504070055 0504070057
0504070060 0506070733 0610079810 0509074846 0403060959 0411068039 0606086093
0606086097 0310057034 0310057035 0606086091 0612080504 0603079025 0609082334
0610079806 0609079889 0609082335 0406061037 0409065507 0409065508 0409065509
0409065510 0409065530 0409065532 0409065534 0409067666 0502072216 0608080023
0608080025 0501072513 0501072514 0611079017 0611081316 0504070053 0607081884
0604081423 0605082514 0601083586 0601083588 0606087507 0611081188 0611081192
0505074492 0607081882 0403059872 0403059874 0404059003 0508081334 0508081336
0502077104 0502077105 0502077107 0612087526 0603079027 0506069871 0601082094
0409064823 0409065992 0503071533 0503071534 0408060133 0403060953 0403060958
0409065529 0413060960 0506069883 0506069884 0509074954 0509074959 0309052386
0312058184 0408060132 0408060134 0601083583 0403060523 0403060525 0604081420
0605082515 0502077103 0503071620 0503071621 0503071622 0401059731 0401059732
0612087534 0507073997 0506070701 0403060952 0502071581 0601083585 0601083590
0401059207 0502072218 0508080997 0509074726 0505074497 0606086099 0606086100
0607081881 0608080024 0502071587 0403060957 0409067665 0604081090 0609082439
0611081315 0402060345 0403059959 0403060524 0601082056 0601082057 0608078598
0310057037 0504070052 0505075447 0502072073 0310057030 0609082438 0501071493
0501071494 0502071887 0503071541 0508081333 0509074845 0407060485 0408067475
0403059951 0403059958 0403060051 0409065994 0409065995 0406062858 0406062859
0502071588 0502071589 0403060053 0601082051 0506070702 0506070704 0312052343
0311050367 0502071585 0502071586 0402060341 0402060344 0406062856 0403060955
0611081189 0611081311 0510072561 0511075671 0502072225 0403060954 0601082055
0402060343 0403060521 0604081416 0604081422 0509074842 0509074843 0509074844
0310057036 0312058183 0612087536 0503071539 0504070058 0506070796 0506070798
0506070800 0612087523 0612087525 0406061036 0406066193 0508081374 0309052272
0509074730 0310057029 0612080502 0604081095 0606087901 0403059953 0405059702
0405059704 0612087682 0612087684 0611079016 0611079019 0611081191 0603079023
0604081094 0606087509 0502072055 0502072056 0502072057 0401059203 0401059734
0408061456 0612080501 0612087531 0510072562 0401059205 0612087524 0509074841
0612087533 0612087535 0511075667 0511075669 0409065531 0409065533 0409065535
0406066191 0406066192 0310057032 0311050368 0401059327 0505074499 0506070681
0601082054 0601083587 0403060054 0403060058 0312058182 0507073995 0507073996
0507074000 0409064961 0409064962 0409064963 0606087903 0606087905 0608078597
0608078599 0609079887 0609079891 0506070734 0404059006 0509074956 0509074960
0408067471 0408067472 0403060052 0609082436 0611079020 0502072072 0502072223
0502072224 0401059330 0405060915 0407058699 0407058700 0504070051 0504070054
0407067635 0501071496 0501072515 0405060917 0406061039 0411068041 0411068201
0312058055 0402060342 0612087532 0312052341 0401059328 0401059329 0402059069
0510072560 0510072565 0403059954 0403059955 0312058181 0606087508 0506070682
0510072564 0407067632 0608078596 0507073998 0509074953 0509074955 0607081885
0608080021 0504070059 0506070732 0506070735 0612087686 0401059733 0401059735
0407060481 0407060484 0409067667 0409067668 0403060056 0403060057 0502072074
0402060434 0405059701 0405060914 0411068202 0411068204 0603079024 0502071584
0502071590 0312052342 0312058185 0309052276 0310057028 0310057033 0512070331
0512070335 0611081187 0611081190 0406059343 0407060482 0408061452 0408067474
0501072516 0502071582 0502071583 0502071889 0502071890 0606087904 0607081894
0409065993 0409065996 0610079804 0610079807 0609082337 0403059957 0403059960
0403060055 0406059342 0406061038 0406066194 0403059873 0403060951 0404059007
0404059008 0611079018 0611081313 0411068203 0604081417 0604081091 0509074951
0509074957 0604081418 0604081419 0604081421 0610079805 0509074724 0509074727
0502071888 0502071891 0605082512 0605082516 0510072563 0512070334 0607081891
0607081895 0506069873 0506069881 0506069882 0506069885 0612087521 0507073999
0611081312 0611081314 0502072222 0503071623 0503071624 0504070056 0505075446
0505075449 0506070731 0605082510 0605082511 0509074721 0508080994 0502077106
0505074494 0505074498 0309052387 0612087683 0604081036 0604081040 0605082513
0601082093 0601082096 0404059001 0404059005 0404059010 0508080991 0508080996
0310057031 0508080992 0508080993 0505074491 0505074500 0504070081 0504070082
0609082332 0612080505 0508081335 0508081373 0612080503 and 0612080506
RECALLING FIRM/MANUFACTURER
Recalling Firm: Sechrist Industries Inc, Anaheim CA, by letters beginning
on November 9, 2007.
Manufacturer: Stryker Medical, Div. of Stryker Corp., Portage, MI. Firm initiated
recall is ongoing.
REASON
Brake Problems: The stretcher brake cams may crack while attempting to lock the
stretcher brake.
VOLUME OF PRODUCT IN COMMERCE
559 units
DISTRIBUTION
Nationwide and Internationally
RECALLS AND FIELD CORRECTIONS: DEVICES - CLASS III
___________________________________
PRODUCT
Biomet 3i, Certain PreFormance Temporary Cylinder, Catalog Number: IPFTC42,
Dental Implant, Recall # Z-0128-2008
CODE
Lot Number: 707303, Expiration Date: 08/31/2012
RECALLING FIRM/MANUFACTURER
Biomet 3i, Palm Beach Gardens, FL, by telephone beginning on September 27,
2007. Firm initiated recall is ongoing.
REASON
Incorrect assembly: The product does not allow for the screw to pass through
the access hole and engage the implant properly. Therefore, the provisional
restoration will not be seated on the implant platform.
VOLUME OF PRODUCT IN COMMERCE
114 units
DISTRIBUTION
Nationwide and Internationally
___________________________________
PRODUCT
Nova 8 Analyzer Calibrator Pack (Catalog/REF # 15196), Recall # Z-0295-2008
CODE
Lot Numbers: 610633, 701034, 704053, 707222
RECALLING FIRM/MANUFACTURER
Nova Biomedical Corp., Waltham, MA, by telephone on September 17, 2007. Firm
initiated recall is complete.
REASON
Elevated Normalized Ionized Calcium (nCa) and Normalized Ionized Magnesium
(nMg) calculated values on patient samples.
VOLUME OF PRODUCT IN COMMERCE
2,411 units
DISTRIBUTION
Nationwide and Internationally