  July 23, 1997                                         97-30
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
  _______________
  PRODUCT      Welsh Farms Ice Cream in 3 gallon bulk
                 containers:  a) Heavenly Hash Ice Cream; 
               b) French Vanilla Ice Cream.  
               Recall #F-513/514-7.
  CODE         a) All date codes prior to 7/19/96 or 20196;
                 b) All date codes prior to 8/5/96 or 21896.
  MANUFACTURER Welsh Farms, Inc., West Caldwell, New Jersey.
  RECALLED BY  Welsh Farms, Inc., Long Valley, New Jersey, by
                 letter dated August 30, 1996.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION New York, New Jersey, Pennsylvania.
  QUANTITY     a) 147 cans were distributed; firm estimated
                 that 100 units remained on market at time of
                 recall initiation; b) 2,315 cans were
                 distributed; firm estimated that 100 units
                 remained on market at time of recall
                 initiation.
  REASON       a) The product does not bear an ingredient
                 statement and contains undeclared egg albumin
                 and almonds.  b) The product does not bear an
                 ingredient statement and contains undeclared
                 egg yolks.
    RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
  _______________
  PRODUCT      No-No brand a) Salsa Diet; b) Sesame Salted
                 Diet Tortilla Corn Flatbread, net weight 4.5
                 ounce. (125.6g cans).  Recall #F-493/494-7.
  CODE         All codes and serial numbers.
  MANUFACTURER Falcone's Cookieland Ltd, Brooklyn, New York.
  RECALLED BY  Wedgie Dietetic Foods, Inc., Brooklyn, New
                 York, by Fax on June 2, 1997.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION New York, New Jersey, Maryland, Pennsylvania,
                 Massachusetts.
  QUANTITY     a) 135 cases; b) 171 cases were distributed.
  REASON       The product contains wheat flour but fails to
                 declare this fact on the ingredient panel.
  
  _______________
  PRODUCT      Bossea brand smoked fish product, vacuum
                 packed and individually packaged in plastic,
                 and were of various weights:
               a) Bossea brand Smoked Mackerel (Whole &
                 Fillet)
               b) Bossea brand Smoked Herring
               c) Bossea brand Smoked Eel
               d) Bossea brand Smoked Whitefish
               e) Bossea brand Smoked Rainbow Trout.
               Recall #F-506/510-7.
  CODE         Smoked Whole Mackerel Code: 2087, date(s)
                 unknown;
               Smoked Mackerel Fillet Code: 2078, date(s)
                 unknown;
               Smoked Herring Code: 2005, date(s) unknown;
               Smoked Eel Code: 2082, date(s) unknown;
               Smoked Whitefish Code: 2081, date(s) unknown;
               Smoked Rainbow Trout Code: 2017, "Dec18 96",
                 "Not To Be Sold Or Consumed After DEC 29,96". 
  MANUFACTURER Bos Smoked Fish, Inc., Woodstock, Ontario,
                 Canada.
  RECALLED BY  Brooklyn Imports Inc., Brooklyn, New York, by
                 visit on. Completed recall resulted from
                 sample analysis and followup by the New York
                 Department of Agriculture and Markets.
  DISTRIBUTION New York, New Jersey, Connecticut, Michigan,
                 Illinois, Virginia, Pennsylvania,
                 Massachusetts.
  QUANTITY     932.6 pound for all five products were
                 distributed.
  REASON       The products contain sodium nitrite, an
                 unapproved food additive.
  
                               -2-_______________
  PRODUCT      Jerusalem World Pure Foods brand Turnip
                 Pickle, a refrigerated, pickled turnip food
                 product in clear glass quart (32 ounce) jar.
               Recall #F-515-7.
  CODE         All product - The firm began coding the
                 product as of June 11, 1997 with the Julian
                 date.
  MANUFACTURER Jerusalem Natural Foods, Inc., Redford,
                 Michigan.
  RECALLED BY  Manufacturer, by telephone or visit by the end
                 of June 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Michigan and Ohio.
  QUANTITY     Approximately 40 cases (12 quarts per case).
  REASON       The product contains D&C Red No. 19, an
                 unapproved color additive.
  
  _______________
  PRODUCT      Keebler Classic Collection French Vanilla
                 Creme Sandwich Cookies, in 12 ounce, 24 ounce,
                 and 300 count boxes.  Recall  #F-516-7.
  CODE         All product manufactured prior to April 9,
                 1997 with shift codes "A", "B" and "C".  The
                 shift code precedes the three digit Julian
                 date code of production for those cookies    
                 made in 1997 (021A made on 1/21/97, first
                 shift), and follows the Julian date code for
                 those cookies made in 1996 (B366 made on
                 12/31/96, second shift).  Product made since
                 April 9, 1997 do not contain the FD&C Yellow 5
                 Lake, and have the shift codes "D", "E" and
                 "F".
  MANUFACTURER Pate Foods, South Beloit, Illinois.
  RECALLED BY  Keebler Company, Elmhurst, Illinois, by
                 memorandum dated April 10, 1997.  Firm-initiated field correction (relabeling)
                 completed.
  DISTRIBUTION Nationwide.
  QUANTITY     31,504 cases were distributed.
  REASON       Product contained undeclared FD&C Yellow No. 5
                 Lake.
  
  _______________
  PRODUCT      No-No brand Diet Tortilla Corn Flatbreads
                 Everything (Sesame-Poppy-Onion-Garlic), net
                 weight 4.5 ounces (125.6g).  Recall #F-517-7.
  CODE         All codes and serial numbers.
  MANUFACTURER Falcone's Cookieland Ltd., Brooklyn, New York.
  RECALLED BY  Wedgie Dietetic Foods, Inc., Brooklyn, New
                 York, by letter on June 24, 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Maryland, New York, Pennsylvania, Minnesota,
                 Massachusetts, Florida.
                               -3-QUANTITY       618 cases (12 units per case) were sold since
                 January 1997.
  REASON       The product contains wheat flour but fails to
                 declare wheat flour on the label.
  
  
  RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
  _______________
  PRODUCT      Hickory Farms Melt Away Mints packed in 16
                 ounce containers:
               a) Hickory Farms Melt Away Mints packed in 16
                 ounce clear plastic bags and in 16 ounce
                 Hickory Farms nostalgic tins
               b) Hickory Farms Mini Melt Away Mints, packed
                 in 16 ounce clear plastic bags.
               Recall #F-511-512-7.
  CODE         All codes are subject to recall - catalog
                 numbers:  regular sized mints in bags - #4750;
                 mini mints in bags - #5959; regular mints
                 packed in nostalgic tins - #7820.
  MANUFACTURER Plantation Candies, Inc., Telford,
                 Pennsylvania.  
  RECALLED BY  Hickory Farms, Inc., Maumee, Ohio, by
                 facsimile on May 20, 1997, followed by E-mail
                 on May 22, 1997.  Firm-initiated recall
                 ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     341,492 containers were distributed.
  REASON       The product contains undeclared certified
                 colors, FD&C Blue No. 1 and FD&C Red No. 3.
  
  _______________
  PRODUCT      Aloe Ginseng Cranberry Beverage, in 1 liter
                 bottles.  Recall #F-518-7.
  CODE         705H02.
  MANUFACTURER Health One's, Brooklyn, New York.
  RECALLED BY  Natural Group, Inc., Modesto, California, by
                 telephone.  Firm-initiated recall completed on
                 or about June 26, 1997.
  DISTRIBUTION California.
  QUANTITY     Firm estimates none remains on the market.
  REASON       The product contained undeclared dextrose.
  
  
  RECALLS AND FIELD CORRECTIONS:  DRUGS --
  CLASS III ==========
  _______________
  PRODUCT      Mexiletine Hydrochloride Capsules, 250 mg, Rx
                 orally active antiarrhythmic used for the
                 treatment of documented life-threatening
                 ventricular arrhythmias, in 100 capsule
                 bottles.  Recall #D-228-7.
  CODE         Lots 100860A, 100861A, 101262A, 104912C.
  
                               -4-MANUFACTURER   Novopharm Ltd., Toronto, Canada.
  RECALLED BY  Novopharm USA, Inc., Schaumburg, Illinois, by
                 letter dated July 9, 1997.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     2,401 bottles were distributed; firm estimated
                 that 25 percent of product remained on market
                 at time of recall initiation.
  REASON       Incorrect labeling -- Some bottles of 250 mg
                 capsules are labeled as containing 150 mg
                 capsules.
  
  
  RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-792/793-7.
  CODE         Unit #2357219.
  MANUFACTURER Central Indiana Regional Blood Center, Inc.,
                 Indianapolis, Indiana.
  RECALLED BY  Manufacturer, by letter dated December 4,
                 1995. Firm-initiated recall completed.
  DISTRIBUTION Indiana and Pennsylvania.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor
                 whose spouse has a history of hepatitis C.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Fresh Frozen Plasma.
               Recall #B-1035/1036-7.
  CODE         Unit #S27628.
  MANUFACTURER Irwin Memorial Blood Centers, San Francisco,
                 California.
  RECALLED BY  Manufacturer, by letter dated March 14, 1996. 
                 Firm-initiated recall completed.
  DISTRIBUTION California.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor who
                 was listed on the State Deferred Donor
                 Registry.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Recovered Plasma.
               Recall #B-1039/1040-7.
  CODE         Unit #41GT71886.
  MANUFACTURER American Red Cross Blood Services, Birmingham,
                 Alabama.
  RECALLED BY  Manufacturer, by letter on or about April 16,
                 1996.  Firm-initiated recall completed.
  DISTRIBUTION Alabama.
  QUANTITY     1 unit of each component was distributed.
  
                               -5-REASON         Blood products tested repeatedly reactive for
                 the hepatitis B surface antigen (HBsAg), and
                 were subjected to additional retesting which
                 was negative.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets.  
               Recall #B-1041/1042-7.
  CODE         Unit #41FY80886.
  MANUFACTURER American Red Cross Blood Services, Birmingham,
                 Alabama.
  RECALLED BY  Manufacturer, by letter on August 28, 1996. 
                 Firm-initiated recall completed.
  DISTRIBUTION Alabama.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a donor who
                 traveled to an area considered endemic for
                 malaria.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c) Recovered
                 Plasma.  Recall #B-1043/1045-7.
  CODE         Unit numbers:  a) 41GM21295, 41LF44653; b)
                 41LF44653; c) 41GM21295, 41LF44653.
  MANUFACTURER American Red Cross Blood Services, Birmingham,
                 Alabama.
  RECALLED BY  Manufacturer, by letters between April 16,
                 1996, and December 12, 1996.  Firm-initiated
                 recall completed.
  DISTRIBUTION Alabama, California.
  QUANTITY     a) 2 units; b) 1 unit; c) 2 units were
                 distributed.
  REASON       Blood products tested negative for the
                 antibody to the human immunodeficiency virus
                 type 1 (anti-HIV-1), but were collected from
                 donors who previously tested repeatedly
                 reactive for anti-HIV-1, Western blot
                 negative.
  
  _______________
  PRODUCT      a) Red Blood Cells; b) Cryoprecipitated AHF;
                 c) Plasma; d) Recovered Plasma.
               Recall #B-1046/1049-7.
  CODE         Unit numbers:  a) 41FV24002, 41GJ37198,
                 41GJ37527, 41GW08651, 41GW08967, 41KC09277,
                 41KC09610; b) 41GJ37198, 41GJ37527;
               c)  41GJ37527; d) 41FV24002, 41GJ37198,
                 41GW08651, 41GW08967, 41KC09277, 41KC09610,
                 41LG53711.
  MANUFACTURER American Red Cross Blood Services, Birmingham,
                 Alabama.
  
                               -6-RECALLED BY    Manufacturer, by letters between April 16,
                 1996, and December 12, 1996.  Firm-initiated
                 recall ongoing.
  DISTRIBUTION Alabama, California, Puerto Rico.
  QUANTITY     a) 7 units; b) 2 units; c) 1 unit; d) 7 units
                 were distributed.
  REASON       Blood products tested negative for the
                 antibody to the hepatitis C virus encoded
                 antigen (anti-HCV), but were collected from
                 donors who previously tested repeatedly
                 reactive for anti-HCV.
  
  _______________
  PRODUCT      "Blood Bank Control System" Software Program
                 Recall #B-1051-7.
  CODE         Software version 4.0.
  MANUFACTURER Blood Bank Computer Systems, Inc., Auburn,
                 Washington.
  RECALLED BY  Manufacturer, by Fax on June 6, 1997.  Firm-initiated field correction completed.
  DISTRIBUTION California, Florida, Illinois, Kansas,
                 Pennsylvania, Texas, Washington state.
  QUANTITY     14 users in the U.S. which have software
                 version 4.0 installed.
  REASON       Computer software contains programming errors
                 which could result in the release of
                 unsuitable blood products.
  
  _______________
  PRODUCT      Source Plasma.  Recall #B-1062-7.
  CODE         Unit #HP99771.
  MANUFACTURER Interstate Blood Bank, Inc., Chicago,
                 Illinois.
  RECALLED BY  Manufacturer, by letter dated May 9, 1997. 
                 Firm-initiated recall completed.
  DISTRIBUTION New Jersey.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor who
                 received a tattoo within 12 months of
                 donation.
  
  _______________
  PRODUCT      Platelets.  Recall #B-1063-7.
  CODE         Unit #29443-7417.
  MANUFACTURER United Blood Services, Chicago, Illinois.
  RECALLED BY  Manufacturer, by telephone on May 13, 1997. 
                 Firm-initiated recall completed.
  DISTRIBUTION Illinois.
  QUANTITY     1 unit was distributed.
  REASON       Platelets were prepared from a donor who had
                 taken aspirin within 3 days of donation.
  
  
                               -7-RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
  _______________
  PRODUCT      Fresh Frozen Plasma.  Recall #B-1001-7.
  CODE         Unit numbers: 4416104, 4422379, 4422395,    
                 4434783   6740379   6740383   7014364
               7014369   7014371   7014372   7014376 
               7074039   7188153   7192601   7195848 
               7229463   7247817   7296027   7351673 
               7369616   7376155   7376914   7378155 
               7386207   7389603   7394208   7396337 
               7581829   7591522   7591546   7635770 
               7638517   7638526   7638533   7638538 
               7638756   7648751   7652746   7652754 
               7655203   7655206   7655210   7665257.
  MANUFACTURER New York Blood Services, also known as New
                 York Blood Center, New York, New York.
  RECALLED BY  Manufacturer, by letter dated January 9, 1997. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION New York and New Jersey.
  QUANTITY     43 units collected from 3/95 to 11/96 were
                 distributed.
  REASON       Blood products contain the preservative Adsol
                 but not labeled as containing that
                 preservative.
   
  _______________
  PRODUCT      a) Red Blood Cells; b) Platelets; c) Recovered
                 Plasma.  Recall #B-1013/1015-7.
  CODE         Unit #1337776.
  MANUFACTURER Central Kentucky Blood Center, Lexington,
                 Kentucky.
  RECALLED BY  Manufacturer, by telephone and letters on
                 December 13, 1996.  Firm-initiated recall
                 completed.
  DISTRIBUTION Kentucky and California.
  QUANTITY     1 unit of each component was distributed.
  REASON       Blood products were collected from a
                 therapeutic donor with hemochromatosis.
  
  _______________
  PRODUCT      Hyp Rho-D Brand Immune Globulin (Human).
               Recall #B-1020-7.
  CODE         Lot #621P15E, 621P15F.
  MANUFACTURER Bayer Corporation, Clayton, North Carolina.
  RECALLED BY  Manufacturer, by letter on June 27, 1997. 
                 Firm-initiated recall completed.
  DISTRIBUTION Nationwide and international.
  QUANTITY     41,190 doses were distributed.
  REASON       Immune Globulin tested subpotent.
  
  _______________
  PRODUCT      Fresh Frozen Plasma.  Recall #B-1034-7.
  CODE         Unit #3099670.
                               -8-MANUFACTURER   Gulf Coast Regional Blood Center, Houston,
                 Texas.
  RECALLED BY  Manufacturer, by Fax on February 10, 1997. 
                 Firm-initiated recall completed.
  DISTRIBUTION Oklahoma.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was not frozen within eight
                 hours of collection.
  
  _______________
  PRODUCT      Recovered Plasma.  Recall #B-1038-7.
  CODE         Unit #41GF45735.
  MANUFACTURER American Red Cross Blood Services, Birmingham,
                 Alabama.
  RECALLED BY  Manufacturer, by letter on June 10, 1996. 
                 Firm-initiated recall completed.
  DISTRIBUTION California.
  QUANTITY     1 unit was distributed.
  REASON       Blood product corresponded to a unit of Red
                 Blood Cells which was implicated in a
                 transfusion reaction.
  
  _______________
  PRODUCT      Red Blood Cells.  Recall #B-1050-7.
  CODE         Unit #49J71353.
  MANUFACTURER American Red Cross Blood Services, Tulsa,
                 Oklahoma.
  RECALLED BY  Manufacturer, by telephone on April 10, 1997. 
                 Firm-initiated recall completed.
  DISTRIBUTION Texas.
  QUANTITY     1 unit was distributed.
  REASON       Blood product was collected from a donor whose
                 health history screening was inadequately
                 performed.
  
  _______________
  PRODUCT      Platelet Pheresis, Leukodepleted.  
               Recall #B-1065-7.
  CODE         Unit #40P52343.
  MANUFACTURER American Red Cross Blood Services, Peoria,
                 Illinois.
  RECALLED BY  Manufacturer, by telephone and fax on March
                 12, 1997.  Firm-initiated recall completed.
  DISTRIBUTION Illinois.
  QUANTITY     1 unit was distributed.
  REASON       Unlicensed blood product was distributed
                 without obliteration of U.S. license.
  
  
                               -9-RECALLS AND FIELD CORRECTIONS:  DEVICES --
  CLASS II =========
  _______________
  PRODUCT      Non-Fiber Optic High Speed Dental Handpieces:
               a) Catalog No. 750044, Tradition Non-Fiber 
                 Optic Handpiece   with conventional chuck (with
                 wrench)
               b) Catalog No. 780044, Traditional L Non-fiber
                 Optic Handpiece   with Power Lever chuck
                 (wrenchless).  Recall Z-674/675-7.
  CODE         All with date code E127 through E227.
  MANUFACTURER Midwest Dental Products (Dentsply), Des
                 Plaines, Illinois.
  RECALLED BY  Manufacturer, by letter dated May 29, 1997. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and China.
  QUANTITY     444 handpieces were distributed.
  REASON       Some of the hand pieces may have loose fitting
                 water spray tubes which may fall out during
                 use, and be swallowed or aspirated.  
  
  _______________
  PRODUCT      Advantage EMG System, Model A100, an
                 electromyogram device used to measure and
                 record electrical activity associated with
                 skeletal muscle.  Recall #Z-687-7.
  CODE         All serial numbers.
  MANUFACTURER Advantage Medical, London, Ontario, Canada.
  RECALLED BY  Manufacturer, by letter dated May 12, 1997. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and Canada.
  QUANTITY     89 units were distributed.
  REASON       The device could cause an electrical burn to
                 the patient.
  
  _______________
  PRODUCT      GDXI Anatomically Correct Lifeline Patient
                 Cable w/Leadwires, Model No. 14081 - Type A,
                 an anatomically correct Lifeline Patient Cable
                 to a Siemens 15 Pin Connector with lead wires,
                 for ECG monitoring in general practitioner
                 settings.  Recall #Z-726-7.
  CODE         Lot numbers of the cable/leadwire: C02105CH,
                 C02504CH, and C02505CH.  NOTE: The finished
                 device kit is NOT identified with lot     
                 numbers.  
  MANUFACTURER Graphic Controls Corporation, Cherry Hill, New
                 Jersey.
  RECALLED BY  Graphic Controls Corporation, Buffalo, New
                 York, by letter dated May 16, 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and international.
  QUANTITY     301 sets were distributed.
  
                               -10-REASON         The lead wires were placed incorrectly into
                 the cable, which caused the lead location to
                 be mislabeled.
  
  _______________
  PRODUCT      Cathcor Remote 20 Inch  Monitor, used with
                 Siemens Cathcor System, an ECG recording
                 device.  Recall #Z-740-7.
  CODE         Part numbers 61-61-942 and 08-99-019.  All
                 units.
  MANUFACTURER Siemens-Elema AB, Life Support Systems
                 Division, Solna, Sweden.
  RECALLED BY  Siemens Medical Systems, Inc., Danvers,
                 Massachusetts, by letter on November 5, 1996. 
                 Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     Undetermined.
  REASON       The monitor used in a cardiac catheterization
                 disconnected from the base pedestal and fell,
                 causing injury.
  
  
  RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
  _______________
  PRODUCT      Clinitek Atlas Reagent Pak, Product #5015,
                 used with Clinitek Atlas Automated Urine
                 Chemistry Analyzer.  Recall #Z-727-7.
  CODE         Lot #1A044017.
  MANUFACTURER Bayer Corporation, Elkhart, Indiana.
  RECALLED BY  Manufacturer, by telephone on May 9, 1997,
                 followed by letter.  Firm-initiated recall
                 completed.
  DISTRIBUTION Nationwide and Puerto Rico.
  QUANTITY     768 units were distributed.
  REASON       The ketone reagent pad is missing in some of
                 the strips in this lot.
  
  _______________
  PRODUCT      1.0mm TI Burr Hole Cover, Model #420.900, used
                 to cover a hole drilled into skull bone so as
                 to result in a flush surface with no recesses
                 or indentations and give a better cosmetic
                 appearance.  Recall #Z-737-7.
  CODE         Lot #A3FN421.
  MANUFACTURER Synthes (USA), Monument, Colorado.
  RECALLED BY  Synthes (USA), Paoli, Pennsylvania, by E-mail
                 April 23, 1996.  Firm-initiated recall
                 completed.
  DISTRIBUTION Nationwide and Canada.
  QUANTITY     80 units were distributed; firm estimates none
                 remains on the market.
  REASON       The 1.0MM Burr Hole Cover was labeled as a
                 1.5MM Burr Hole Cover.
                               -11-_______________
  PRODUCT      TDx/TDxFLx Estriol Reagent Pack, List No.
                 9112-60, used with either Total Estriol
                 Calibrators (lists 9112-01) or Free Estriol
                 Calibrators (list 9118-01) in a Fluorescence
                 Polarization Immunoassay (FPIA) in-vitro
                 estriol or free estriol in human serum, plasma
                 or urine.  Recall #Z-738-7.
  CODE         Lot Nos. 22109Q100, 23635Q100, 25589Q100, and
                 26075Q100.
  MANUFACTURER Abbott Health Products, Inc., Barceloneta,
                 Puerto Rico.
  RECALLED BY  Abbott Laboratories, Diagnostics Division,
                 Abbott Park, Illinois, by letter on July 2,
                 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION California, Kentucky, Georgia, Illinois,
                 Indiana, Louisiana, Maryland, West Virginia,
                 Ohio, Pennsylvania, Texas, and international.
  QUANTITY     1,226 reagents packs were distributed; firm
                 estimated that 10 percent remained on market
                 at time of recall initiation.
  REASON       Storage of the antibody S reagent for extended
                 time at elevated temperatures can cause heat
                 stressing which results in depressed
                 millipolarization values and display of the
                 printed error code "PO Too Small" during
                 calibration.
  
  _______________
  PRODUCT      Quidel QuickVue Chlamydia Test, Catalog No.
                 B006.  Recall #Z-739-7.
  CODE         Lot Nos. B006 J04501, B006 J04801, B006
                 J04802, B006 J06901, B006 J08001, B006 J09001,
                 B006 J10601, and B006 J11201.
  MANUFACTURER Quidel Corporation, San Diego, California.
  RECALLED BY  Manufacturer, by telephone and by letter dated
                 May 21, 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide and international.
  QUANTITY     1,809 (30 tests/kit) kits were distributed.
  REASON       Faint or no results are obtained with the
                 positive control solution.
  
  _______________
  PRODUCT      Siemens 12 Lead ECG Patient Cable and Cable
                 Kit, used with the Siemens Cathcor Equipment
                 an ECG recording device:
               a) 12 Lead ECG Patient Cable, Part No.
                 0899300.
               b) 12 Lead ECG Cable Kit, Part No. 0899399.
               Recall #Z-741/742-7.
  CODE         Lot #R0823.  
  MANUFACTURER ConMed Corporation, Utica, New York.
  
                               -12-RECALLED BY    Siemens Medical Systems, Inc., Danvers,
                 Massachusetts, by letters on May 23, 1997, and
                 June 5, 1997.  Firm-initiated recall ongoing.
  DISTRIBUTION Nationwide.
  QUANTITY     184 cables were distributed.
  REASON       The indicator tabs for the Left Arm (LA) and
                 Left Leg (LL) leads are reversed.
  
                             -13-
  
  END OF ENFORCEMENT FOR JULY 23, 1997.  BLANK PAGES MAY FOLLOW.

                                   ####