RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II
=========================
_______________
PRODUCT        Morphine Sulfate Suppositories, 20 mg, Rx analgesic, in
               boxes of 12.  Recall #D-187-5.
CODE           5B680 and 4L619.
MANUFACTURER   Paddock Laboratories, Inc., Minneapolis, Minnesota.
RECALLED BY    Manufacturer, by letter May 23, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,506 boxes were distributed.
REASON         Potency is incorrectly declared as 10 mg on the back panel
               of the carton.  The principal panel and a side panel are
               correctly labeled as 20 mg.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III
========================
_______________
PRODUCT        Sinucon Syrup, an OTC antihistamine/decongestant, under the
               following labels:  LuChem Pharmaceuticals/Sinucon Syrup, in
               1 pint and 1 gallon bottles, Major Pharmaceuticals/Nalgest
               Syrup, in 1 pint bottles.  Recall #D-186-5.
CODE           Lot #7986.
MANUFACTURER   H.N. Norton Company, Shreveport, Louisiana.  (Note:  Firm
               name change to Zenith Goldline Shreveport, Inc.)
RECALLED BY    Manufacturer, by telephone and by letter April 12, 1995. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       1,463 1-pint bottles (LuChem label), 88 1-gallon bottles
               (LuChem label), 712 1-pint bottles (Major label) were
               distributed.
REASON         Potential for crystal formation when product is exposed to
               low temperatures.

_______________
PRODUCT        Physostigmine Sulfate Ophthalmic Ointment (Eserine Sulfate
               1/4%), 1/4 ounce, USP, in 1/8 ounce (3.5g) tubes, Rx
               anticholinergic compound.  Recall #D-188-5.
CODE           Lot #2537 EXP 4/96.
MANUFACTURER   Altana, Inc., Melville, New York.
RECALLED BY    Manufacturer, by letter June 2, 1995.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       9,360 tubes were distributed.
REASON         Potency not assured through expiration date.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II
=====================
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Recovered Plasma. 
               Recall #B-365/367-5.
CODE           Unit #BUI 17033-9420.
MANUFACTURER   Blood Systems, Inc., doing business as United Blood
               Services, McAllen, Texas.
RECALLED BY    Manufacturer, by letter January 16, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Texas, Florida.
QUANTITY       1 unit of each component.
REASON         Blood products, which tested negative for the antibody to
               immunodeficiency virus type 1 (anti-HIV-1), but were
               collected from a donor who previously tested repeatedly for
               anti-HIV-1, were distributed.

_______________
PRODUCT        (a) Red Blood Cells; (b) Red Blood Cells, for further
               manufacture; (c) Platelets; (d) Platelets, for further
               manufacture; (e) Fresh Frozen Plasma; (f) Recovered Plasma. 
               Recall #B-369/374-5.
CODE           All components manufactured from one unit of whole blood are
               identified with the same unit number, date drawn and blood
               type.
               PRODUCT - Red Cells 
               UNIT NUMBER       DATE DRAWN 
               32-01692           9/19/88 
               80-13522          12/13/88 
               80-49257           4/18/89 
               80-65054           6/23/89 
               80-76848           8/15/89 
               80-93517          11/27/89 
               81-07207          12/12/89 

                                    -2-               81-29213           4/17/90 
               81-32562           3/23/90 
               81-56337           6/22/90 
               81-68130           8/15/90 
               81-90159           12/7/90 
               81-93272          12/11/90 
               82-30275           4/30/91 
               82-42819           6/28/91 
               82-62865           9/20/91 
               83-58293           11/9/92 
               83-66304           12/7/92 
               83-70551          12/18/92 
               83-95722           4/26/93 
               84-11704           6/18/93 
               84-20663            8/5/93 
               84-21055            8/9/93 
               84-22774            9/9/93 
               84-35813          11/15/93 
               84-44998           11/8/93 
               84-47126          12/10/93 
               84-48827           12/9/93 
               84-54928            2/9/94 
               84-70094           3/11/94 
               PRODUCT - Platelets 
               UNIT NUMBER       DATE DRAWN 
               80-02515           11/2/88 
               80-13522          12/13/88 
               80-49257           4/18/89 
               80-76848           8/15/89 
               81-07207          12/12/89 
               81-32562           3/23/90 
               81-56337           6/22/90 
               81-68130           8/15/90 
               81-90159           12/7/90 
               81-93272          12/11/90 
               82-30275           4/30/91 
               82-42819           6/28/91 
               82-62865           9/20/91 
               83-70551          12/18/92 
               83-95722           4/26/93 
               84-11704           6/18/93 
               84-20663            8/5/93 
               84-21055            8/9/93 
               84-35813          11/15/93 
               84-44998           11/8/93 
               84-47126          12/10/93 
               84-48827           12/9/93 
               84-54928            2/9/94 
               PRODUCT - Pooled Platelets 
               UNIT NUMBER    DATE DRAWN 
               84-48827          12/9/93 

                                    -3-               PRODUCT - Fresh Frozen Pooled 
               UNIT NUMBER    DATE DRAWN 
               81-32562          3/23/90 
               81-56337          6/22/90 
               81-90159          12/7/90 
               82-30275          4/30/91 
               83-70551         12/18/92 
               84-22774           9/9/93 
               PRODUCT - Plasma 
               UNIT NUMBER    DATE DRAWN 
               82-42819          6/28/91 
               82-62865          9/20/91 
               83-58293          11/9/92 
               83-66304          12/7/92 
               83-95722          4/26/93 
               84-11704          6/18/93 
               84-20663           8/5/93 
               84-21055           8/9/93 
               84-35813         11/15/93 
               84-44998          11/8/93 
               84-47126         12/10/93 
               84-48827          12/9/93 
               84-54928           2/9/94 
               84-70094          3/11/94 
MANUFACTURER   The Blood Center for Southeast Louisiana, New Orleans,
               Louisiana.
RECALLED BY    Manufacturer, by telephone, fax, and letter June 14, 1994. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Louisiana,  New York, California, Florida, North Carolina.
QUANTITY       73 units.
REASON         Blood products, which tested negative for hepatitis B
               surface antigen (HBsAg), but were collected from donors who
               previously tested repeatedly reactive for HBsAg, were
               distributed.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III
====================
_______________
PRODUCT        Recovered Plasma.  Recall #B-381-5.
CODE           Unit #26029-5282.
MANUFACTURER   Blood Systems, Inc., doing business as United Blood Services
               (UBS), Fort Smith Arkansas.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by fax October 6,
               1992.  Firm-initiated recall complete.
DISTRIBUTION   North Carolina.
QUANTITY       1 unit.
REASON         Blood product, associated with a unit of Red Blood Cells
               which was contaminated with Propionibacterium acnes, was
               distributed. 

                                    -4-_______________
PRODUCT        Reagent Red Blood Cells RESOLVE(tm) Panel A.
               Recall #B-382-5.
CODE           KA810 EXP 7/26/94.
MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
RECALLED BY    Manufacturer, by letter July 18, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       648 packages.
REASON         Reagent Red Blood Cell panels which contained errors on the
               extended antigen profile, were distributed.

_______________
PRODUCT        Reagent Red Blood Cells, RESOLVE(tm) Panel A.
               Recall #B-383-5.
CODE           Lot RA834 EXP 1/17/95.
MANUFACTURER   Ortho Diagnostic Systems, Inc., Raritan, New Jersey.
RECALLED BY    Manufacturer, by fax December 29, 1994, and by telephone
               between February 27, 1995 and March 20, 1995, and by letters
               of December 30, 1994 and March 23, 1995.  Firm-initiated
               recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       802 packages.
REASON         Lea positive Reagent Red Blood Cells, incorrectly labeled as
               Lea negative, were distributed.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II
=======================
_______________
PRODUCT        Transformer used with Ultrasonic Scaler used for cleaning
               teeth, Model Turbo 25/30.  Recall #Z-821-5.
CODE           Serial numbers of Turbo Scaler #44867 - 45190. The
               transformer does not have a serial number.
MANUFACTURER   Woet Tsern Electronic Company, Ltd., Taipei, Taiwan
               (contract manufacturer).
RECALLED BY    Parkell Products, Inc., Farmingdale, New York, by telephone
               August 12, 1994.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       326 units were distributed; firm estimates none remains on
               market. 
REASON         The transformer was manufactured incorrectly which may lead
               to lack of grounding which could cause the device to short
               out.

_______________
PRODUCT        Single Lumen Femoral Catheters under the Acuflex and Neostar
               label.  Recall #Z-822-5.
CODE           Lot numbers:   AT157, AT158, AT159, AT199, AT200, AT222,
               AT223, AT225, AT245, AT246, AR260, AS197, AS198, AS244,
               AS245, AS246, AS247, AS347, AT156.
MANUFACTURER   Neostar Medical Technologies, Inc., King of Prussia,
               Pennsylvania.

                                    -5-RECALLED BY    Manufacturer, by letter April 27, 1994 and May 3, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   New York, West Virginia, Missouri, Colorado, England, Egypt,
               India, Germany, Malaysia, Pakistan.
QUANTITY       6,089 units (kits and single catheters) were distributed;
               firm estimates none remains on the market
REASON         A 0.038" guidewire would not allow passage of the catheter's
               tip.

_______________
PRODUCT        Baxter Arterial and Venous Bloodlines:
               (a) Baxter Arterial Blood Tubing Set for Hemodialysis,
               Product Code 5M4253M;
               (b) Baxter Arterial Blood Tubing Set for Hemodialysis with
               Male Luer Rotating Lock, Product Code 5M4262M;
               (c) Baxter Arterial Blood Tubing Set for Hemodialysis with
               Male Luer Rotating Lock and Pressure Pillow Tubing, Product
               Code 5M4283M;
               (d) Baxter Venous Blood Tubing Set for Hemodialysis with two
               lengths of tubing, Product Code 5M4491M;
               (e) Baxter Venous Blood Tubing Set for Hemodialysis with
               four lengths of tubing, Product Code 5M4484M.
               Recall #Z-831/835-5.
CODE           All lots.
MANUFACTURER   Productos Urologos de Mexico S.A. de C.V., Mexicali-Baja,
               Mexico.
RECALLED BY    Baxter Healthcare Corporation, McGaw Park, Illinois, by
               letter May 16, 1995.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide, Brazil, Venezuela.
QUANTITY       121,000 sets were distributed; firm estimated that 25
               percent of product remained on market at time of recall
               initiation.
REASON         Blood tubing might not completely set in the dialyzer
               connector resulting in a potential for blood leakage.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III
======================
_______________
PRODUCT        2991 Triple Processing Set, tubing set used in bone marrow
               procedures.  Recall #Z-779-5.
CODE           Catalog #912647-901, Lot #07Z15125.
MANUFACTURER   Cobe BCT, Inc., Lakewood, Colorado.
RECALLED BY    Manufacturer, by letter October 17, 1994.  Firm-initiated
               recall complete.
DISTRIBUTION   Massachusetts, Colorado, New York, Nebraska, Tennessee,
               Texas, Belgium, France, Spain, England, Israel.
QUANTITY       216 units were distributed; firm estimates none remains on
               the market.

                                    -6-REASON         A misassembly error resulted in the set containing an
               incorrect tubing size than intended.  The affected tubing is
               used at the peristaltic set during a bone marrow procedure. 
               The failure to occlude the tubing results in the fluid being
               unable to flow as directed by the pump which can result in
               procedural inefficiency.

_______________
PRODUCT        Preference 4-Pack, BC 8.7, DIA 14.4, Extended Wear Contact
               Lenses.  Recall #Z-828-5.
CODE           L09483 (on outer carton).
MANUFACTURER   Coopervision, Inc., Scottsville, New York.
RECALLED BY    Manufacturer, by telephone April 7, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Arkansas, California, Florida, Georgia, Illinois,
               Massachusetts, Missouri, North Carolina, New Jersey, New
               York, Oklahoma, Oregon, Pennsylvania, Tennessee, Texas,
               Wisconsin, West Virginia.
QUANTITY       23 4-packs were distributed.  As of May 17, 1995 4 4-packs
               remained unaccounted for.
REASON         The lenses are labeled with the wrong power.

_______________
PRODUCT        Cooperclear, BC 8.7, DIA 14.4, Daily Wear Contact Lenses,
               sold in individual vials.  Recall #Z-829-5.
CODE           Lot #L09538.
MANUFACTURER   Coopervision, Inc., Scottsville, New York.
RECALLED BY    Manufacturer, by telephone April 21, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Arkansas, California, Florida, Georgia, Illinois,
               Massachusetts, Missouri, North Carolina, New Jersey, New
               York, Oklahoma, Oregon, Pennsylvania, Tennessee, Texas,
               Wisconsin, West Virginia.
QUANTITY       26 lenses were distributed.  As of May 9, 1995, 11 lenses
               remained unaccounted for.
REASON         The lenses are labeled with the wrong power.

_______________
PRODUCT        Diamedix brand Anti-DNA Microassay, Catalog #783-200, 96
               test set, for in-vitro diagnostic use.  Recall #Z-836-5.
CODE           Lot #S0404A EXP 8/95.
MANUFACTURER   Diamedix Corporation, Miami, Florida.
RECALLED BY    Manufacturer, by letter February 10, 1995.  Firm-initiated
               recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       459 kits were distributed.
REASON         Some of the wells in Anti-DNA Microassay, may not have been
               coated with antigen during the manufacturing process.


                                    -7-_______________
PRODUCT        FIRST Choice Reagent Strips for use with Glucometer II
               (GII), Glucometer II with Memory, and Glucometer M Blood
               Glucose Meters: (a) Catalog No. 91050; (b) Catalog No.
               91051; (c) FIRST Choice Reagent Strips for use with DIASCAN
               Meter, Catalog No. 91080.  Recall #Z-837/839-5. 
CODE           Lot Numbers:  (a) F51445, F51446, F51447, F51449, F51450,
               F51451; (b) F51450, F51451, F51454; (c) F70983.
MANUFACTURER   Polymer Technology International (PTI), Issaquah,
               Washington.
RECALLED BY    Manufacturer, by letter March 5, 1995 and April 14, 1995,
               followed by telephone the week of May 1, 1995.  Firm-
               initiated recall ongoing.
DISTRIBUTION   Nationwide, Belgium, Taiwan, Poland, Japan, Pakistan,
               Lebanon.
QUANTITY       103,327 packages were distributed.
REASON         The blood glucose strips exceeded their specifications which
               could provide inaccurate patient results.